MS
Martyn Smith
Fri, Dec 14, 2007 8:03 PM
Hi,
I saw the original message for the requirement of a portable reference
needed to maintain an accuracy of 100 Hz at 47 GHz I think.
I know I'm a bit late, but there has been a lot of misguided advice, in my
opinion.
If the user truly needs 100 Hz at 47 GHz, that is an accuracy of 2E-9.
Now, I supply the most stable oscillators from Wenzel and other suppliers as
well as making my own oscillators.
There's no way of achieving an accuracy of 2E-9 unless you go to a rubidium
oscillator. At least not without a massive amount of work.
Although many crystal oscillators quote specs of 1E-10/day or even
1E-11/day, this is only achieved after the crystal has been running for a
long time. A long time meaning weeks or months.
For example, Wenzel quote 1E-10/day but only after 30 days continuous use.
Before that they drift by about 5E-8/day. They need about 3 to six months
before they start to age slowly. Also an OXCO's retrace is usually very
poor. Typically 1E-7 to 1E-8. Achieving 2E-9 accuracy is near on
impossible for an OXCO.
However, a rubidium oscillator will achieve this in minutes. My own range
of rubidium oscillators has typically re-trace of 1E-10 in less than 10
minutes.
Also my rubidium's have excellent Allan Variance (<1E-11 in 1 sec for my
units), equaling most OXCO'S oscillators, so there's no need to lock a OXCO
to the rubidium, that was also suggested.
Simply use a good quality rubidium.
With some manufacturers selling rubidium oscillators for under $1000, the
cost is not as much as what would be needed to keep an OXCO.
Best Regards
Martyn
This Email is from:
Martyn Smith
Precision Test Systems LTD
Tel: +44 (0) 1245 329608
Email: martyn@ptsyst.com
Web: www.ptsyst.com
NOTICE - This message contains legally privileged and confidential
information intended only for the use of the addressee named above. If you
are not the intended recipient of this message, you are hereby notified that
you must not disseminate, distribute, copy or take any action in reliance on
it. If you have received this message in error, please notify Precision Test
Systems LTD.
This communication represents the originator's personal views and opinions,
which do not necessarily reflect those of Precision Test Systems Ltd. from
whom it emanates. Further, Precision Test Systems Ltd. will not be held
responsible for its contents and/or attachments, if the communication
contravenes the rules of Precision Test Systems Ltd's email policy.
----- Original Message -----
From: time-nuts-request@febo.com
To: time-nuts@febo.com
Sent: Tuesday, December 11, 2007 11:22 PM
Subject: time-nuts Digest, Vol 41, Issue 35
Send time-nuts mailing list submissions to
time-nuts@febo.com
To subscribe or unsubscribe via the World Wide Web, visit
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
or, via email, send a message with subject or body 'help' to
time-nuts-request@febo.com
You can reach the person managing the list at
time-nuts-owner@febo.com
When replying, please edit your Subject line so it is more specific
than "Re: Contents of time-nuts digest..."
Today's Topics:
- Re: need recomendation for a portable 10mhz reference
oscilator (Magnus Danielson)
- Re: is there a "best bet" advanced hobbyist buildable
GPSDOdesign? (Bruce Griffiths)
- Re: need recomendation for a portable 10mhz reference
oscilator (John Ackermann N8UR)
- Re: need recomendation for a portable 10mhz reference
oscilator (Bruce Griffiths)
- Re: need recomendation for a portable 10mhz reference
oscilator (Bruce Griffiths)
- Re: Fury Interface Board simulation results (Bruce Griffiths)
- Re: need recomendation for a portable 10mhz reference
oscilator (Hal Murray)
- Re: need recomendation for a portable 10mhz reference
oscilator (Alan Melia)
Message: 1
Date: Tue, 11 Dec 2007 21:52:55 +0100 (CET)
From: Magnus Danielson magnus@rubidium.dyndns.org
Subject: Re: [time-nuts] need recomendation for a portable 10mhz
reference oscilator
To: time-nuts@febo.com, eric.fort@gmail.com
Message-ID: 20071211.215255.-743699540.cfmd@bredband.net
Content-Type: Text/Plain; charset=us-ascii
From: "Eric Fort" eric.fort@gmail.com
Subject: [time-nuts] need recomendation for a portable 10mhz reference
oscilator
Date: Tue, 11 Dec 2007 11:38:40 -0800
Message-ID: 2ad2af430712111138q23eb4dfm76d9a138d76365fc@mail.gmail.com
Eric,
I'm looking for a fairly basic, relatively simple 10 Mhz PORTABLE
reference (probably quartz based) with enough stability over a period
of a week of outdoor temp extremes to keep a 47Ghz transmitter locked
within 100hz while mountaintop contesting. Suggestions are
appreciated. something that is small and ran off 12 volts dc (car
battery) would be a definite plus.
Have a look at the Oscilloquartz 8711/8712 (G option, you would need
< 3E-10/day, so 2E-10/day fits your needs, but maybe you can live with the
164,5 Hz the standard 5E-10/day would give you) for instance.
Hook up a DAC and a small PIC or AVR to keep the DAC easilly controlable
when
calibrating the frequency in the home lab.
There are many similar oscillators which will do the same trick. A 10811
would
also do the same thing, but it requires a higher voltage but a switcher
would
double up the 12V.
Cheers,
Magnus
Message: 2
Date: Wed, 12 Dec 2007 09:53:36 +1300
From: Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] is there a "best bet" advanced hobbyist
buildable GPSDOdesign?
To: Tom Van Baak tvb@leapsecond.com, Discussion of precise time and
frequency measurement time-nuts@febo.com
Message-ID: 475EF8D0.7030400@xtra.co.nz
Content-Type: text/plain; charset=ISO-8859-1
Tom Van Baak wrote:
Hi,
Like many, I've acquired a fair amount of surplus test equipment off of
Ebay
which could use the services of good master frequency standard. So I'm
looking to discipline an HP 10811 VXCO to provide this.
Any general consensus about the best design for a hobbyist to build?
I'm familiar with the Brooks Shera design, the G4JNT Jupiter-T design,
the TAC-2 circuit, and the VE2ZAZ design. I take it from discussions
I've seen in the archives of this list that the VE2ZAZ design makes a
number of simplification/performance tradeoffs.
Is there a design I haven't listed which is "better" than the others?
I'm quite familiar with microcontrollers, FPGAs, spinning my own
PCBs, etc, so I'll roll my own if I have to, but I'd prefer to build
a variation on someone's tried and true design.
I'm aware of products like the Fury, but I'd like something I could
tinker
with, and the cost is hard to justify for a hobbyist.
Scott
Tom
What about the short term performance?
Its relatively easy to achieve a stability of 1E-13 for an averaging
time of 1 day, achieving good short or medium term stability is more
difficult.
If you want simplicity and higher performance you can do far better with
fewer parts,
An expensive high resolution DAC can be replaced with a software
sigma-delta DAC that has higher resolution.
The complex phase detector can be replaced with a D flipflop.
Add a microprocessor plus an opamp or 2 to filter and scale the EFC
voltage and thats about all thats required in addition to a good GPS
timing receiver.
For improved performance a hardware circuit to correct the PPS sawtooth
error will improve the medium term stability significantly when using a
high performance GPS timing receiver that provides an estimate of this
error.
Both the Brooks Shera and the James Miller designs have inadequate phase
error measurement resolution to achieve good short and medium term
stability.
However, this is only noticeable when using high performance GPS timing
receivers (M12+T, M12MT etc) and a high quality OCXO (10811A etc).
Bruce
Message: 3
Date: Tue, 11 Dec 2007 15:56:42 -0500
From: John Ackermann N8UR jra@febo.com
Subject: Re: [time-nuts] need recomendation for a portable 10mhz
reference oscilator
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Message-ID: 475EF98A.6010207@febo.com
Content-Type: text/plain; charset=ISO-8859-1; format=flowed
Hal Murray wrote:
I'm picturing an open hill top at night with clear skies so there would
be
lots of radiation cooling. Of course, with an open hill top, GPS
recption
would be easy.
As great as GPSDO are, the problem with using them in a portable
environment is the time it takes to do a survey and then get to final
lock; mountain-toppers may not be in one place long enough.
One thought I've had is to use a cheap Rubidium for long term stability,
and phase lock a good OCXO to it for short term stability and phase
noise (ie, use a loop time constant of perhaps 100 seconds). That would
make a portable package that would meet both the short and long term
stability requirements.
John
Message: 4
Date: Wed, 12 Dec 2007 10:05:49 +1300
From: Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] need recomendation for a portable 10mhz
reference oscilator
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Message-ID: 475EFBAD.7000708@xtra.co.nz
Content-Type: text/plain; charset=ISO-8859-1
Hal Murray wrote:
A good (and well-aged) double-oven OCXO with aging compensation should
be able to provide that type of performance. The problem is: how
much physical movement will the unit experience? What are the
temperature extremes that are expected, and how fast does the
ambient move between the temp extremes? Can it be GPS locked during
that time?
Assuming you don't GPS lock it...
How well will two units from the same manufacturer/batch track?
I'm assuming they will see roughly the same temperature profile and that
you
start with them running at the same frequency.
Has anybody made a triple oven setup? I'm thinking of a big box with a
few
holes and a temperature controlled fan. The idea is to let the oven (and
electronics) keep itself warm when the ambient temperature gets very
cold.
I'm picturing an open hill top at night with clear skies so there would
be
lots of radiation cooling. Of course, with an open hill top, GPS
recption
would be easy.
Hal,
A triple oven is a particularly bad idea, its just too difficult to
achieve a wide operating temperature range without losing temperature
control or cooking everything in the innermost oven unless one uses a
watercooled peltier device to control the outer oven temperature.
A well designed single oven with feedforward (or the outer oven of a
dual oven system if you cant redesign/rebuild the inner oven) can easily
achieve a thermal gain of more than 1E5 see:
http://tf.nist.gov/timefreq/general/pdf/595.pdf
A bootstrapped oven design like that used by Wenzel (not as new as he'd
have you believe - its been in use in one form or another for the better
part of a century) will also help in improving performance. The optimum
solution is of course to combine the 2 techniques.
http://www.wenzel.com/documents/Sub-pico%20Multiplier.pdf
With a thermal gain of 1E5 for the outer oven temperature fluctuations
at the outer shell of the inner oven will only be a 1 millikelvin or so
(assuming the ambient temperature range is less than 100K).
Bruce
Message: 5
Date: Wed, 12 Dec 2007 10:11:51 +1300
From: Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] need recomendation for a portable 10mhz
reference oscilator
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Message-ID: 475EFD17.3070500@xtra.co.nz
Content-Type: text/plain; charset=ISO-8859-1
John Ackermann N8UR wrote:
I'm picturing an open hill top at night with clear skies so there would
be
lots of radiation cooling. Of course, with an open hill top, GPS
recption
would be easy.
As great as GPSDO are, the problem with using them in a portable
environment is the time it takes to do a survey and then get to final
lock; mountain-toppers may not be in one place long enough.
One thought I've had is to use a cheap Rubidium for long term stability,
and phase lock a good OCXO to it for short term stability and phase
noise (ie, use a loop time constant of perhaps 100 seconds). That would
make a portable package that would meet both the short and long term
stability requirements.
John
John
Why not lock it to a GPS timing receiver output when its at home, and
allow the controller to learn the aging and tempco characteristics etc
of the oscillator being disciplined and use this information to correct
for temperature variations etc at the remote site?
The model used can be as complex as required to achieve the desired
performance and a PC can be used to derive optimum correction
coefficients, the micro only has to perform the simpler task of applying
these corrections.
Bruce
Message: 6
Date: Wed, 12 Dec 2007 11:10:38 +1300
From: Bruce Griffiths bruce.griffiths@xtra.co.nz
Subject: Re: [time-nuts] Fury Interface Board simulation results
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Message-ID: 475F0ADE.1050505@xtra.co.nz
Content-Type: text/plain; charset=ISO-8859-1
xaos wrote:
George
In the JFET frequency doubler:
- Surely the 50 ohms should be is series with the the voltage source V3
for the simulation?
Correct. I did a cut and paste "without rotate" there.
- Usually a 1:4 impedance ratio step up transformer on the input is
about right with a similar transformer used to step down the output
(perhaps an even higher impedance ratio transformer (8:1, 9:1?? may be
better). The maximum turns ratio depends somewhat on the maximum
allowable drain voltage swing which in turn is limited by the drain
supply voltage.
I will re-run with the this in mind and will repost.
The idea is that the FETs are not operated in the square law region but
are actually switched off alternately so that the combined drain current
waveform is a rectified sinewave with a small dc offset.
With a well balanced circuit the odd harmonic amplitudes (including the
fundamental) at the output will be small with the 20MHz 4th harmonic of
the fundamental being the largest component.
Achieving low distortion isnt the aim, but achieving low phase noise is.
Any residual harmonics and subharmonics can be removed with tuned
circuit traps (dont use a high Q bandpass filter as it will inevitably
have a high phase shift tempco and may add flicker phase noise depending
on the components used). A relatively broadband low pass filter
contributes little phase shift (and associated flicker phase noise as
well as little phase shift tempco) at 10MHz as do a series of relatively
high Q LC traps. High performance is rarely achieved by the
simplest/obvious solution.
If you have difficulty finding suitable JFETs a high impedance bipolar
junction transistor (BJT) implementation is also possible however
biasing is a little trickier and requires a couple of extra BJTs to
ensure reasonable thermal stability of the bias currents.
In the voltage offset circuit:
- The junction of R9 and R12 should be connected to the offset source
(+5V??).
This is what I didn't get about the original circuit. I was scratching
my head as to the purpose of a variable R9 when
the only value that made sense was when it was set to the "0" position.
I dont understand what the 7812 does in this circuit.
It was there in the original circuit so I modeled accordingly. Actually
I ran some simulations with PS voltage variation
to see the effect as well. In that case I removed the 7812.
Isolation amplifier looks OK.
Bruce
This is a very nice design and the AC analysis shows a bandpass around
10 MHz with 20 MHz BW.
The only issue I see there is that it needs 50 parts of 6 different
values.
This might be difficult to implement on a small board and would be prone
to error for people trying to
assemble their own board.
The way I see the final product is a bare board and possibly a kit. It
has to be easy to assemble and minimize
errors.
Actually, I was looking at the isolation amplifier provided in the C.M.
Felton Paper:
http://www.darksmile.net/ee/Superimposing_Low-Phase-Noise_Low-Drift_Instrumentation_Techniques_On_RF_Design.pdf
It is relatively simple and would use similar devices as the Frequency
doubler.
The trouble with this design is that it has a low input impedance and
thus is unsuitable for OCXOs that require a high impedance load (eg
10544A).
Its gain isnt adjustable and its difficult to obtain suitable dual JFETS
(you would have to use a couple of J310's to achieve similar performance
since U431 isnt readily obtainable, the LS840 is definitely unsuitable
as its minimum Idss isnt high enough ), it also has relatively low
reverse isolation. Simplicity isnt the key to high performance.
I will upload more info as I compile it.
George
As great as GPSDO are, the problem with using them in a portable
environment is the time it takes to do a survey and then get to final
lock; mountain-toppers may not be in one place long enough.
What if you use a navigation GPS unit rather than a timing unit?
How long does a survey take (in good conditions) and what sort of time
constant would you want to use with a navigation box?
Does anybody have graphs of the PPS wander from a navigation unit?
I like the lock-to-rubidium suggestion.
--
These are my opinions, not necessarily my employer's. I hate spam.
Message: 8
Date: Tue, 11 Dec 2007 23:21:47 -0000
From: "Alan Melia" alan.melia@btinternet.com
Subject: Re: [time-nuts] need recomendation for a portable 10mhz
reference oscilator
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Message-ID: 009301c83c4c$9be64680$0900a8c0@Lark
Content-Type: text/plain; charset="iso-8859-1"
Hi John, Andy Talbot G4JNT developed the a Jupiter based GPSDO for just
this
purpose, several are used by hill-toppers in the UK.
http://www.scrbg.org/g4jnt/freqlock.htm
The idea was that it would settle and be accurate in the time it took to
seet up the microwave part of the station.
I hope that helps/
Alan G3NYK
----- Original Message -----
From: "John Ackermann N8UR" jra@febo.com
To: "Discussion of precise time and frequency measurement"
time-nuts@febo.com
Sent: Tuesday, December 11, 2007 8:56 PM
Subject: Re: [time-nuts] need recomendation for a portable 10mhz reference
oscilator
I'm picturing an open hill top at night with clear skies so there would
lots of radiation cooling. Of course, with an open hill top, GPS
As great as GPSDO are, the problem with using them in a portable
environment is the time it takes to do a survey and then get to final
lock; mountain-toppers may not be in one place long enough.
One thought I've had is to use a cheap Rubidium for long term stability,
and phase lock a good OCXO to it for short term stability and phase
noise (ie, use a loop time constant of perhaps 100 seconds). That would
make a portable package that would meet both the short and long term
stability requirements.
John
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
and follow the instructions there.
Hi,
I saw the original message for the requirement of a portable reference
needed to maintain an accuracy of 100 Hz at 47 GHz I think.
I know I'm a bit late, but there has been a lot of misguided advice, in my
opinion.
If the user truly needs 100 Hz at 47 GHz, that is an accuracy of 2E-9.
Now, I supply the most stable oscillators from Wenzel and other suppliers as
well as making my own oscillators.
There's no way of achieving an accuracy of 2E-9 unless you go to a rubidium
oscillator. At least not without a massive amount of work.
Although many crystal oscillators quote specs of 1E-10/day or even
1E-11/day, this is only achieved after the crystal has been running for a
long time. A long time meaning weeks or months.
For example, Wenzel quote 1E-10/day but only after 30 days continuous use.
Before that they drift by about 5E-8/day. They need about 3 to six months
before they start to age slowly. Also an OXCO's retrace is usually very
poor. Typically 1E-7 to 1E-8. Achieving 2E-9 accuracy is near on
impossible for an OXCO.
However, a rubidium oscillator will achieve this in minutes. My own range
of rubidium oscillators has typically re-trace of 1E-10 in less than 10
minutes.
Also my rubidium's have excellent Allan Variance (<1E-11 in 1 sec for my
units), equaling most OXCO'S oscillators, so there's no need to lock a OXCO
to the rubidium, that was also suggested.
Simply use a good quality rubidium.
With some manufacturers selling rubidium oscillators for under $1000, the
cost is not as much as what would be needed to keep an OXCO.
Best Regards
Martyn
This Email is from:
Martyn Smith
Precision Test Systems LTD
Tel: +44 (0) 1245 329608
Email: martyn@ptsyst.com
Web: www.ptsyst.com
NOTICE - This message contains legally privileged and confidential
information intended only for the use of the addressee named above. If you
are not the intended recipient of this message, you are hereby notified that
you must not disseminate, distribute, copy or take any action in reliance on
it. If you have received this message in error, please notify Precision Test
Systems LTD.
This communication represents the originator's personal views and opinions,
which do not necessarily reflect those of Precision Test Systems Ltd. from
whom it emanates. Further, Precision Test Systems Ltd. will not be held
responsible for its contents and/or attachments, if the communication
contravenes the rules of Precision Test Systems Ltd's email policy.
----- Original Message -----
From: <time-nuts-request@febo.com>
To: <time-nuts@febo.com>
Sent: Tuesday, December 11, 2007 11:22 PM
Subject: time-nuts Digest, Vol 41, Issue 35
> Send time-nuts mailing list submissions to
> time-nuts@febo.com
>
> To subscribe or unsubscribe via the World Wide Web, visit
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> or, via email, send a message with subject or body 'help' to
> time-nuts-request@febo.com
>
> You can reach the person managing the list at
> time-nuts-owner@febo.com
>
> When replying, please edit your Subject line so it is more specific
> than "Re: Contents of time-nuts digest..."
>
>
> Today's Topics:
>
> 1. Re: need recomendation for a portable 10mhz reference
> oscilator (Magnus Danielson)
> 2. Re: is there a "best bet" advanced hobbyist buildable
> GPSDOdesign? (Bruce Griffiths)
> 3. Re: need recomendation for a portable 10mhz reference
> oscilator (John Ackermann N8UR)
> 4. Re: need recomendation for a portable 10mhz reference
> oscilator (Bruce Griffiths)
> 5. Re: need recomendation for a portable 10mhz reference
> oscilator (Bruce Griffiths)
> 6. Re: Fury Interface Board simulation results (Bruce Griffiths)
> 7. Re: need recomendation for a portable 10mhz reference
> oscilator (Hal Murray)
> 8. Re: need recomendation for a portable 10mhz reference
> oscilator (Alan Melia)
>
>
> ----------------------------------------------------------------------
>
> Message: 1
> Date: Tue, 11 Dec 2007 21:52:55 +0100 (CET)
> From: Magnus Danielson <magnus@rubidium.dyndns.org>
> Subject: Re: [time-nuts] need recomendation for a portable 10mhz
> reference oscilator
> To: time-nuts@febo.com, eric.fort@gmail.com
> Message-ID: <20071211.215255.-743699540.cfmd@bredband.net>
> Content-Type: Text/Plain; charset=us-ascii
>
> From: "Eric Fort" <eric.fort@gmail.com>
> Subject: [time-nuts] need recomendation for a portable 10mhz reference
> oscilator
> Date: Tue, 11 Dec 2007 11:38:40 -0800
> Message-ID: <2ad2af430712111138q23eb4dfm76d9a138d76365fc@mail.gmail.com>
>
> Eric,
>
>> I'm looking for a fairly basic, relatively simple 10 Mhz PORTABLE
>> reference (probably quartz based) with enough stability over a period
>> of a week of outdoor temp extremes to keep a 47Ghz transmitter locked
>> within 100hz while mountaintop contesting. Suggestions are
>> appreciated. something that is small and ran off 12 volts dc (car
>> battery) would be a definite plus.
>
> Have a look at the Oscilloquartz 8711/8712 (G option, you would need
> < 3E-10/day, so 2E-10/day fits your needs, but maybe you can live with the
> 164,5 Hz the standard 5E-10/day would give you) for instance.
> Hook up a DAC and a small PIC or AVR to keep the DAC easilly controlable
> when
> calibrating the frequency in the home lab.
>
> There are many similar oscillators which will do the same trick. A 10811
> would
> also do the same thing, but it requires a higher voltage but a switcher
> would
> double up the 12V.
>
> Cheers,
> Magnus
>
>
>
> ------------------------------
>
> Message: 2
> Date: Wed, 12 Dec 2007 09:53:36 +1300
> From: Bruce Griffiths <bruce.griffiths@xtra.co.nz>
> Subject: Re: [time-nuts] is there a "best bet" advanced hobbyist
> buildable GPSDOdesign?
> To: Tom Van Baak <tvb@leapsecond.com>, Discussion of precise time and
> frequency measurement <time-nuts@febo.com>
> Message-ID: <475EF8D0.7030400@xtra.co.nz>
> Content-Type: text/plain; charset=ISO-8859-1
>
> Tom Van Baak wrote:
>>> Hi,
>>>
>>> Like many, I've acquired a fair amount of surplus test equipment off of
>>> Ebay
>>> which could use the services of good master frequency standard. So I'm
>>> looking to discipline an HP 10811 VXCO to provide this.
>>>
>>> Any general consensus about the best design for a hobbyist to build?
>>> I'm familiar with the Brooks Shera design, the G4JNT Jupiter-T design,
>>> the TAC-2 circuit, and the VE2ZAZ design. I take it from discussions
>>> I've seen in the archives of this list that the VE2ZAZ design makes a
>>> number of simplification/performance tradeoffs.
>>>
>>> Is there a design I haven't listed which is "better" than the others?
>>> I'm quite familiar with microcontrollers, FPGAs, spinning my own
>>> PCBs, etc, so I'll roll my own if I have to, but I'd prefer to build
>>> a variation on someone's tried and true design.
>>>
>>> I'm aware of products like the Fury, but I'd like something I could
>>> tinker
>>> with, and the cost is hard to justify for a hobbyist.
>>>
>>> Scott
>>>
>>
>> Scott,
>>
>> Hard to say which is better at this point; there are a number
>> of variables, not the least of which is the intrinsic short-term
>> stability of the OCXO you use.
>>
>> Do have a close look at James Miller's GPSDO:
>>
>> http://www.jrmiller.demon.co.uk/projects/freqstd/frqstd.htm
>>
>> I recently tested one and it makes it to 1e-13 at one day, which
>> is really nice for a simple, cheap, homebrew GPSDO.
>>
>> /tvb
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
> Tom
>
> What about the short term performance?
> Its relatively easy to achieve a stability of 1E-13 for an averaging
> time of 1 day, achieving good short or medium term stability is more
> difficult.
>
> If you want simplicity and higher performance you can do far better with
> fewer parts,
> An expensive high resolution DAC can be replaced with a software
> sigma-delta DAC that has higher resolution.
> The complex phase detector can be replaced with a D flipflop.
> Add a microprocessor plus an opamp or 2 to filter and scale the EFC
> voltage and thats about all thats required in addition to a good GPS
> timing receiver.
> For improved performance a hardware circuit to correct the PPS sawtooth
> error will improve the medium term stability significantly when using a
> high performance GPS timing receiver that provides an estimate of this
> error.
>
> Both the Brooks Shera and the James Miller designs have inadequate phase
> error measurement resolution to achieve good short and medium term
> stability.
> However, this is only noticeable when using high performance GPS timing
> receivers (M12+T, M12MT etc) and a high quality OCXO (10811A etc).
>
> Bruce
>
>
>
> ------------------------------
>
> Message: 3
> Date: Tue, 11 Dec 2007 15:56:42 -0500
> From: John Ackermann N8UR <jra@febo.com>
> Subject: Re: [time-nuts] need recomendation for a portable 10mhz
> reference oscilator
> To: Discussion of precise time and frequency measurement
> <time-nuts@febo.com>
> Message-ID: <475EF98A.6010207@febo.com>
> Content-Type: text/plain; charset=ISO-8859-1; format=flowed
>
> Hal Murray wrote:
>> I'm picturing an open hill top at night with clear skies so there would
>> be
>> lots of radiation cooling. Of course, with an open hill top, GPS
>> recption
>> would be easy.
>
> As great as GPSDO are, the problem with using them in a portable
> environment is the time it takes to do a survey and then get to final
> lock; mountain-toppers may not be in one place long enough.
>
> One thought I've had is to use a cheap Rubidium for long term stability,
> and phase lock a good OCXO to it for short term stability and phase
> noise (ie, use a loop time constant of perhaps 100 seconds). That would
> make a portable package that would meet both the short and long term
> stability requirements.
>
> John
>
>
>
> ------------------------------
>
> Message: 4
> Date: Wed, 12 Dec 2007 10:05:49 +1300
> From: Bruce Griffiths <bruce.griffiths@xtra.co.nz>
> Subject: Re: [time-nuts] need recomendation for a portable 10mhz
> reference oscilator
> To: Discussion of precise time and frequency measurement
> <time-nuts@febo.com>
> Message-ID: <475EFBAD.7000708@xtra.co.nz>
> Content-Type: text/plain; charset=ISO-8859-1
>
> Hal Murray wrote:
>>> A good (and well-aged) double-oven OCXO with aging compensation should
>>> be able to provide that type of performance. The problem is: how
>>> much physical movement will the unit experience? What are the
>>> temperature extremes that are expected, and how fast does the
>>> ambient move between the temp extremes? Can it be GPS locked during
>>> that time?
>>>
>>
>> Assuming you don't GPS lock it...
>>
>> How well will two units from the same manufacturer/batch track?
>>
>> I'm assuming they will see roughly the same temperature profile and that
>> you
>> start with them running at the same frequency.
>>
>>
>> Has anybody made a triple oven setup? I'm thinking of a big box with a
>> few
>> holes and a temperature controlled fan. The idea is to let the oven (and
>> electronics) keep itself warm when the ambient temperature gets very
>> cold.
>>
>> I'm picturing an open hill top at night with clear skies so there would
>> be
>> lots of radiation cooling. Of course, with an open hill top, GPS
>> recption
>> would be easy.
>>
>>
>>
>>
> Hal,
>
> A triple oven is a particularly bad idea, its just too difficult to
> achieve a wide operating temperature range without losing temperature
> control or cooking everything in the innermost oven unless one uses a
> watercooled peltier device to control the outer oven temperature.
>
> A well designed single oven with feedforward (or the outer oven of a
> dual oven system if you cant redesign/rebuild the inner oven) can easily
> achieve a thermal gain of more than 1E5 see:
>
> http://tf.nist.gov/timefreq/general/pdf/595.pdf
>
>
>
> A bootstrapped oven design like that used by Wenzel (not as new as he'd
> have you believe - its been in use in one form or another for the better
> part of a century) will also help in improving performance. The optimum
> solution is of course to combine the 2 techniques.
>
> http://www.wenzel.com/documents/Sub-pico%20Multiplier.pdf
>
> With a thermal gain of 1E5 for the outer oven temperature fluctuations
> at the outer shell of the inner oven will only be a 1 millikelvin or so
> (assuming the ambient temperature range is less than 100K).
>
> Bruce
>
>
>
> ------------------------------
>
> Message: 5
> Date: Wed, 12 Dec 2007 10:11:51 +1300
> From: Bruce Griffiths <bruce.griffiths@xtra.co.nz>
> Subject: Re: [time-nuts] need recomendation for a portable 10mhz
> reference oscilator
> To: Discussion of precise time and frequency measurement
> <time-nuts@febo.com>
> Message-ID: <475EFD17.3070500@xtra.co.nz>
> Content-Type: text/plain; charset=ISO-8859-1
>
> John Ackermann N8UR wrote:
>> Hal Murray wrote:
>>
>>> I'm picturing an open hill top at night with clear skies so there would
>>> be
>>> lots of radiation cooling. Of course, with an open hill top, GPS
>>> recption
>>> would be easy.
>>>
>>
>> As great as GPSDO are, the problem with using them in a portable
>> environment is the time it takes to do a survey and then get to final
>> lock; mountain-toppers may not be in one place long enough.
>>
>> One thought I've had is to use a cheap Rubidium for long term stability,
>> and phase lock a good OCXO to it for short term stability and phase
>> noise (ie, use a loop time constant of perhaps 100 seconds). That would
>> make a portable package that would meet both the short and long term
>> stability requirements.
>>
>> John
>>
> John
>
> Why not lock it to a GPS timing receiver output when its at home, and
> allow the controller to learn the aging and tempco characteristics etc
> of the oscillator being disciplined and use this information to correct
> for temperature variations etc at the remote site?
>
> The model used can be as complex as required to achieve the desired
> performance and a PC can be used to derive optimum correction
> coefficients, the micro only has to perform the simpler task of applying
> these corrections.
>
> Bruce
>
>
>
> ------------------------------
>
> Message: 6
> Date: Wed, 12 Dec 2007 11:10:38 +1300
> From: Bruce Griffiths <bruce.griffiths@xtra.co.nz>
> Subject: Re: [time-nuts] Fury Interface Board simulation results
> To: Discussion of precise time and frequency measurement
> <time-nuts@febo.com>
> Message-ID: <475F0ADE.1050505@xtra.co.nz>
> Content-Type: text/plain; charset=ISO-8859-1
>
> xaos wrote:
>> Bruce Griffiths wrote:
>>
>>> George
>>>
>>> In the JFET frequency doubler:
>>> 1) Surely the 50 ohms should be is series with the the voltage source V3
>>> for the simulation?
>>>
>>>
>> Correct. I did a cut and paste "without rotate" there.
>>
>>> 2) Usually a 1:4 impedance ratio step up transformer on the input is
>>> about right with a similar transformer used to step down the output
>>> (perhaps an even higher impedance ratio transformer (8:1, 9:1?? may be
>>> better). The maximum turns ratio depends somewhat on the maximum
>>> allowable drain voltage swing which in turn is limited by the drain
>>> supply voltage.
>>>
>>>
>> I will re-run with the this in mind and will repost.
>>
> The idea is that the FETs are not operated in the square law region but
> are actually switched off alternately so that the combined drain current
> waveform is a rectified sinewave with a small dc offset.
> With a well balanced circuit the odd harmonic amplitudes (including the
> fundamental) at the output will be small with the 20MHz 4th harmonic of
> the fundamental being the largest component.
> Achieving low distortion isnt the aim, but achieving low phase noise is.
> Any residual harmonics and subharmonics can be removed with tuned
> circuit traps (dont use a high Q bandpass filter as it will inevitably
> have a high phase shift tempco and may add flicker phase noise depending
> on the components used). A relatively broadband low pass filter
> contributes little phase shift (and associated flicker phase noise as
> well as little phase shift tempco) at 10MHz as do a series of relatively
> high Q LC traps. High performance is rarely achieved by the
> simplest/obvious solution.
> If you have difficulty finding suitable JFETs a high impedance bipolar
> junction transistor (BJT) implementation is also possible however
> biasing is a little trickier and requires a couple of extra BJTs to
> ensure reasonable thermal stability of the bias currents.
>>> In the voltage offset circuit:
>>>
>>> 1) The junction of R9 and R12 should be connected to the offset source
>>> (+5V??).
>>>
>>>
>> This is what I didn't get about the original circuit. I was scratching
>> my head as to the purpose of a variable R9 when
>> the only value that made sense was when it was set to the "0" position.
>>
>>
>>> I dont understand what the 7812 does in this circuit.
>>>
>>>
>> It was there in the original circuit so I modeled accordingly. Actually
>> I ran some simulations with PS voltage variation
>> to see the effect as well. In that case I removed the 7812.
>>
>>> Isolation amplifier looks OK.
>>>
>>> Bruce
>>>
>>>
>> This is a very nice design and the AC analysis shows a bandpass around
>> 10 MHz with 20 MHz BW.
>>
>> The only issue I see there is that it needs 50 parts of 6 different
>> values.
>>
>> This might be difficult to implement on a small board and would be prone
>> to error for people trying to
>> assemble their own board.
>>
>> The way I see the final product is a bare board and possibly a kit. It
>> has to be easy to assemble and minimize
>> errors.
>>
>> Actually, I was looking at the isolation amplifier provided in the C.M.
>> Felton Paper:
>>
>> http://www.darksmile.net/ee/Superimposing_Low-Phase-Noise_Low-Drift_Instrumentation_Techniques_On_RF_Design.pdf
>>
>> It is relatively simple and would use similar devices as the Frequency
>> doubler.
>>
> The trouble with this design is that it has a low input impedance and
> thus is unsuitable for OCXOs that require a high impedance load (eg
> 10544A).
> Its gain isnt adjustable and its difficult to obtain suitable dual JFETS
> (you would have to use a couple of J310's to achieve similar performance
> since U431 isnt readily obtainable, the LS840 is definitely unsuitable
> as its minimum Idss isnt high enough ), it also has relatively low
> reverse isolation. Simplicity isnt the key to high performance.
>> I will upload more info as I compile it.
>>
>> George
>>
>
>
>
>
> ------------------------------
>
> Message: 7
> Date: Tue, 11 Dec 2007 14:57:38 -0800
> From: Hal Murray <hmurray@megapathdsl.net>
> Subject: Re: [time-nuts] need recomendation for a portable 10mhz
> reference oscilator
> To: Discussion of precise time and frequency measurement
> <time-nuts@febo.com>
> Message-ID: <20071211225739.00A28BE31@ip-64-139-1-69.sjc.megapath.net>
> Content-Type: text/plain; charset=us-ascii
>
>
>> As great as GPSDO are, the problem with using them in a portable
>> environment is the time it takes to do a survey and then get to final
>> lock; mountain-toppers may not be in one place long enough.
>
> What if you use a navigation GPS unit rather than a timing unit?
>
> How long does a survey take (in good conditions) and what sort of time
> constant would you want to use with a navigation box?
>
> Does anybody have graphs of the PPS wander from a navigation unit?
>
>
> I like the lock-to-rubidium suggestion.
>
>
> --
> These are my opinions, not necessarily my employer's. I hate spam.
>
>
>
>
>
>
> ------------------------------
>
> Message: 8
> Date: Tue, 11 Dec 2007 23:21:47 -0000
> From: "Alan Melia" <alan.melia@btinternet.com>
> Subject: Re: [time-nuts] need recomendation for a portable 10mhz
> reference oscilator
> To: "Discussion of precise time and frequency measurement"
> <time-nuts@febo.com>
> Message-ID: <009301c83c4c$9be64680$0900a8c0@Lark>
> Content-Type: text/plain; charset="iso-8859-1"
>
> Hi John, Andy Talbot G4JNT developed the a Jupiter based GPSDO for just
> this
> purpose, several are used by hill-toppers in the UK.
> http://www.scrbg.org/g4jnt/freqlock.htm
> The idea was that it would settle and be accurate in the time it took to
> seet up the microwave part of the station.
>
> I hope that helps/
> Alan G3NYK
>
> ----- Original Message -----
> From: "John Ackermann N8UR" <jra@febo.com>
> To: "Discussion of precise time and frequency measurement"
> <time-nuts@febo.com>
> Sent: Tuesday, December 11, 2007 8:56 PM
> Subject: Re: [time-nuts] need recomendation for a portable 10mhz reference
> oscilator
>
>
>> Hal Murray wrote:
>> > I'm picturing an open hill top at night with clear skies so there would
> be
>> > lots of radiation cooling. Of course, with an open hill top, GPS
> recption
>> > would be easy.
>>
>> As great as GPSDO are, the problem with using them in a portable
>> environment is the time it takes to do a survey and then get to final
>> lock; mountain-toppers may not be in one place long enough.
>>
>> One thought I've had is to use a cheap Rubidium for long term stability,
>> and phase lock a good OCXO to it for short term stability and phase
>> noise (ie, use a loop time constant of perhaps 100 seconds). That would
>> make a portable package that would meet both the short and long term
>> stability requirements.
>>
>> John
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
>
>
>
> ------------------------------
>
> _______________________________________________
> time-nuts mailing list
> time-nuts@febo.com
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
> End of time-nuts Digest, Vol 41, Issue 35
> *****************************************
>