My first guess would be that the zero pot is dirty, with intermittent
contact. Try spritzing it with a cleaner -- I use LPS1 and it works
great on my old Fluke gear. Some pots are sealed like the old AB pots
ude in Tek and Fluke and HP stuff -- I just drill a small hole through
the back and use the sprizer nozzle to get the cleaner in -- works wonders.
Also, when i redid my 845, I found that even though the electrolytics
checked OK, they still needed to be replaced with verifiably low ESR units.
Dick Moore
My friend asked help with his Fluke 845A. It is late colour, very
clean inside and outside. The higher ranges work fine and the readings
are accurate (for example when the needle is on the 1V scale mark
multimeter reads 1.002V). So the chopper seems to be working.
The problem is instability on the lowest ranges. Noise level is ok,
around 200nVpp on the 1uV range, but zeroing is extremely difficult.
Even a tiny adjustment results a huge change with delay.
Amplifier DC voltages were as expected, but we found out that
adjusting the ZERO potentiometer affects the +/-15V power supply
rails. The overall supply voltage changes or the ground shifts or
both. For example the negative supply can vary from -11 to -18V. Any
change in supply/ground changes the zero setting and we have a chain
reaction.
All the electrolytics measured like new. The +/-15V switching power
supply works fine when we replaced the rest of the circuit with load
resistors. Because of the global feedback it is very difficult to
divide the circuit into parts and check each of them individually.
Any help much appreciated.
My first guess would be that the zero pot is dirty, with intermittent
contact. Try spritzing it with a cleaner -- I use LPS1 and it works
great on my old Fluke gear. Some pots are sealed like the old AB pots
ude in Tek and Fluke and HP stuff -- I just drill a small hole through
the back and use the sprizer nozzle to get the cleaner in -- works wonders.
Also, when i redid my 845, I found that even though the electrolytics
checked OK, they still needed to be replaced with verifiably low ESR units.
Dick Moore
> My friend asked help with his Fluke 845A. It is late colour, very
> clean inside and outside. The higher ranges work fine and the readings
> are accurate (for example when the needle is on the 1V scale mark
> multimeter reads 1.002V). So the chopper seems to be working.
>
> The problem is instability on the lowest ranges. Noise level is ok,
> around 200nVpp on the 1uV range, but zeroing is extremely difficult.
> Even a tiny adjustment results a huge change with delay.
>
> Amplifier DC voltages were as expected, but we found out that
> adjusting the ZERO potentiometer affects the +/-15V power supply
> rails. The overall supply voltage changes or the ground shifts or
> both. For example the negative supply can vary from -11 to -18V. Any
> change in supply/ground changes the zero setting and we have a chain
> reaction.
>
> All the electrolytics measured like new. The +/-15V switching power
> supply works fine when we replaced the rest of the circuit with load
> resistors. Because of the global feedback it is very difficult to
> divide the circuit into parts and check each of them individually.
>
> Any help much appreciated.