volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

HP 3456A AC cal

RM
Richard Moore
Thu, Sep 11, 2014 11:30 PM

I'm trying to get my head around the AC cal procedure for the HP 3456A
-- there is no full-scale cal, just some extremely finicky tweaks for
the bottom ends of three of the four ranges. I really don't understand
how this is supposed to work, especially since it is very accurate on
all DC ranges (up to 500V -- above that some heating of the divider R
seems to occur and it loses accuracy as the voltage increases). But on
AC, the tweaks don't result in anything like an accurate level reading
compare to my 3458A -- at 10VAC it is off by hundreds of ppm. Anybody
shed any light on this? Is there a way to get full scale cal points in
AC mode? Any mods to fix this -- lovely meter otherwise; I much prefer
it to the 34401A I used to have, except for the overall bigness....

I'm trying to get my head around the AC cal procedure for the HP 3456A -- there is no full-scale cal, just some extremely finicky tweaks for the bottom ends of three of the four ranges. I really don't understand how this is supposed to work, especially since it is very accurate on all DC ranges (up to 500V -- above that some heating of the divider R seems to occur and it loses accuracy as the voltage increases). But on AC, the tweaks don't result in anything like an accurate level reading compare to my 3458A -- at 10VAC it is off by hundreds of ppm. Anybody shed any light on this? Is there a way to get full scale cal points in AC mode? Any mods to fix this -- lovely meter otherwise; I much prefer it to the 34401A I used to have, except for the overall bigness....
MK
M K
Fri, Sep 12, 2014 7:04 AM

On 12/09/2014 00:30, Richard Moore wrote:

I'm trying to get my head around the AC cal procedure for the HP 3456A
-- there is no full-scale cal, just some extremely finicky tweaks for
the bottom ends of three of the four ranges. I really don't understand
how this is supposed to work, especially since it is very accurate on
all DC ranges (up to 500V -- above that some heating of the divider R
seems to occur and it loses accuracy as the voltage increases). But on
AC, the tweaks don't result in anything like an accurate level reading
compare to my 3458A -- at 10VAC it is off by hundreds of ppm. Anybody
shed any light on this? Is there a way to get full scale cal points in
AC mode? Any mods to fix this -- lovely meter otherwise; I much prefer
it to the 34401A I used to have, except for the overall bigness....

Hi Richard,

Does the meter have a DC to RMS converter chip in it, because if it does
you have to account for the self noise and offset of the gilbert cell at
the heart of the converter.

On 12/09/2014 00:30, Richard Moore wrote: > I'm trying to get my head around the AC cal procedure for the HP 3456A > -- there is no full-scale cal, just some extremely finicky tweaks for > the bottom ends of three of the four ranges. I really don't understand > how this is supposed to work, especially since it is very accurate on > all DC ranges (up to 500V -- above that some heating of the divider R > seems to occur and it loses accuracy as the voltage increases). But on > AC, the tweaks don't result in anything like an accurate level reading > compare to my 3458A -- at 10VAC it is off by hundreds of ppm. Anybody > shed any light on this? Is there a way to get full scale cal points in > AC mode? Any mods to fix this -- lovely meter otherwise; I much prefer > it to the 34401A I used to have, except for the overall bigness.... > Hi Richard, Does the meter have a DC to RMS converter chip in it, because if it does you have to account for the self noise and offset of the gilbert cell at the heart of the converter.
JP
John Phillips
Fri, Sep 12, 2014 7:36 AM

From what I remember: The 3455A has a good AC front end but is a real pain

to calibrate with a lot of adjustment (read flexibility). I think HP gave
up a lot of accuracy for a simpler calibration. The 3457A seems to do AC a
lot better than the 3456A. Nether meter come close to doing what the 3458A
does.

On Fri, Sep 12, 2014 at 12:04 AM, M K m1k3k1@hotmail.com wrote:

On 12/09/2014 00:30, Richard Moore wrote:

I'm trying to get my head around the AC cal procedure for the HP 3456A --
there is no full-scale cal, just some extremely finicky tweaks for the
bottom ends of three of the four ranges. I really don't understand how this
is supposed to work, especially since it is very accurate on all DC ranges
(up to 500V -- above that some heating of the divider R seems to occur and
it loses accuracy as the voltage increases). But on AC, the tweaks don't
result in anything like an accurate level reading compare to my 3458A -- at
10VAC it is off by hundreds of ppm. Anybody shed any light on this? Is
there a way to get full scale cal points in AC mode? Any mods to fix this
-- lovely meter otherwise; I much prefer it to the 34401A I used to have,
except for the overall bigness....

Hi Richard,

Does the meter have a DC to RMS converter chip in it, because if it does
you have to account for the self noise and offset of the gilbert cell at
the heart of the converter.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.

--

John Phillips

>From what I remember: The 3455A has a good AC front end but is a real pain to calibrate with a lot of adjustment (read flexibility). I think HP gave up a lot of accuracy for a simpler calibration. The 3457A seems to do AC a lot better than the 3456A. Nether meter come close to doing what the 3458A does. On Fri, Sep 12, 2014 at 12:04 AM, M K <m1k3k1@hotmail.com> wrote: > On 12/09/2014 00:30, Richard Moore wrote: > >> I'm trying to get my head around the AC cal procedure for the HP 3456A -- >> there is no full-scale cal, just some extremely finicky tweaks for the >> bottom ends of three of the four ranges. I really don't understand how this >> is supposed to work, especially since it is very accurate on all DC ranges >> (up to 500V -- above that some heating of the divider R seems to occur and >> it loses accuracy as the voltage increases). But on AC, the tweaks don't >> result in anything like an accurate level reading compare to my 3458A -- at >> 10VAC it is off by hundreds of ppm. Anybody shed any light on this? Is >> there a way to get full scale cal points in AC mode? Any mods to fix this >> -- lovely meter otherwise; I much prefer it to the 34401A I used to have, >> except for the overall bigness.... >> >> Hi Richard, > > Does the meter have a DC to RMS converter chip in it, because if it does > you have to account for the self noise and offset of the gilbert cell at > the heart of the converter. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/volt-nuts > and follow the instructions there. > -- *John Phillips*
MK
M K
Fri, Sep 12, 2014 8:28 PM

On 12/09/2014 08:36, John Phillips wrote:

From what I remember: The 3455A has a good AC front end but is a real pain
to calibrate with a lot of adjustment (read flexibility). I think HP gave
up a lot of accuracy for a simpler calibration. The 3457A seems to do AC a
lot better than the 3456A. Nether meter come close to doing what the 3458A
does.

On Fri, Sep 12, 2014 at 12:04 AM, M K m1k3k1@hotmail.com wrote:

On 12/09/2014 00:30, Richard Moore wrote:

I'm trying to get my head around the AC cal procedure for the HP 3456A --
there is no full-scale cal, just some extremely finicky tweaks for the
bottom ends of three of the four ranges. I really don't understand how this
is supposed to work, especially since it is very accurate on all DC ranges
(up to 500V -- above that some heating of the divider R seems to occur and
it loses accuracy as the voltage increases). But on AC, the tweaks don't
result in anything like an accurate level reading compare to my 3458A -- at
10VAC it is off by hundreds of ppm. Anybody shed any light on this? Is
there a way to get full scale cal points in AC mode? Any mods to fix this
-- lovely meter otherwise; I much prefer it to the 34401A I used to have,
except for the overall bigness....

Hi Richard,

Does the meter have a DC to RMS converter chip in it, because if it does
you have to account for the self noise and offset of the gilbert cell at
the heart of the converter.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.

I downloaded the service maual from the BAMA archive, it is an absolute
value circuit followed by a logging anp, not quite the AD634 gilbert
cell, but still prone to the same offset and noise creating a
non-linearity at the bottom of each range.

On 12/09/2014 08:36, John Phillips wrote: > From what I remember: The 3455A has a good AC front end but is a real pain > to calibrate with a lot of adjustment (read flexibility). I think HP gave > up a lot of accuracy for a simpler calibration. The 3457A seems to do AC a > lot better than the 3456A. Nether meter come close to doing what the 3458A > does. > > On Fri, Sep 12, 2014 at 12:04 AM, M K <m1k3k1@hotmail.com> wrote: > >> On 12/09/2014 00:30, Richard Moore wrote: >> >>> I'm trying to get my head around the AC cal procedure for the HP 3456A -- >>> there is no full-scale cal, just some extremely finicky tweaks for the >>> bottom ends of three of the four ranges. I really don't understand how this >>> is supposed to work, especially since it is very accurate on all DC ranges >>> (up to 500V -- above that some heating of the divider R seems to occur and >>> it loses accuracy as the voltage increases). But on AC, the tweaks don't >>> result in anything like an accurate level reading compare to my 3458A -- at >>> 10VAC it is off by hundreds of ppm. Anybody shed any light on this? Is >>> there a way to get full scale cal points in AC mode? Any mods to fix this >>> -- lovely meter otherwise; I much prefer it to the 34401A I used to have, >>> except for the overall bigness.... >>> >>> Hi Richard, >> Does the meter have a DC to RMS converter chip in it, because if it does >> you have to account for the self noise and offset of the gilbert cell at >> the heart of the converter. >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/volt-nuts >> and follow the instructions there. >> > > I downloaded the service maual from the BAMA archive, it is an absolute value circuit followed by a logging anp, not quite the AD634 gilbert cell, but still prone to the same offset and noise creating a non-linearity at the bottom of each range.