Re: [time-nuts] Frequency standards for different tau in Allen Dev measurement

TK
Taka Kamiya
Fri, Feb 21, 2020 10:26 PM

I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point?

1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle.

2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly?

3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway.

4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here. 

5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough?
I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics.

(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote:  

I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab)

It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term.
Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.
That being the case, why this consensus of what time source to use for what tau?
I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME.
I'm still trying to wrap my small head around this. 


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point? 1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle. 2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly? 3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway. 4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here.  5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough? I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics. --------------------------------------- (Mr.) Taka Kamiya KB4EMF / ex JF2DKG On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote: I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab) It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term. Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s. That being the case, why this consensus of what time source to use for what tau? I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME. I'm still trying to wrap my small head around this.  --------------------------------------- (Mr.) Taka Kamiya KB4EMF / ex JF2DKG _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow the instructions there.
BK
Bob kb8tq
Fri, Feb 21, 2020 10:43 PM

HI

On Feb 21, 2020, at 5:26 PM, Taka Kamiya via time-nuts time-nuts@lists.febo.com wrote:

I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point?

  1. A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle.

It does not come up with a 1000 s tau ADEV by doing that. The gate time would need to be open for 1000 seconds.

  1. I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly?

If you are measuring the difference between two one 1 pps signals, and they are < 0.1 seconds apart, and the TICC has a resolution of 0.1 ns ….

Then the reference clock for the TICC only needs to be good to 0.1 ns / 0.1 s = 1x10^-9

  1. I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,  Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway.

A counter that will do 1x10^-12 at 1 second would need errors below 1 ps. I do not know of any such counter. As you go to other tau’s (like 1,000 seconds)
a counter that will keep a gate open for 1000 seconds might do the trick.

  1. Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here.

We have been doing that off list ….

  1. One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough?

Look at the counter’s stated accuracy information and do the math. If it’s a 100 ps counter, then that is 100 x 10^-12 at 1 second. If there are no other errors, and
the reference is perfect, you will hit 10X better each time the gate increases by 10X.

Bob

I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics.

(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote:  

I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab)

It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term.
Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.
That being the case, why this consensus of what time source to use for what tau?
I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME.
I'm still trying to wrap my small head around this.


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

HI > On Feb 21, 2020, at 5:26 PM, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote: > > I'm sorry, I messed up. I jumped on more advance topic than I intended. I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain. I bulletized (is that a word?) the original question with my NEW understanding. Would someone please respond for me, point-to-point? > > 1) A frequency counter that measures DUT basically puts out a reading every second during the measurement. When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s. I understand now, Adev is about phase, not the frequency. But assuming DUT is sine wave, if there is enough phase change, frequency do change. I think of phase change as frequency change that is less than full cycle. So how does counters that outputs every 1 second end up in tau of 1000s? It will entirely miss phase change that spans more than 1 cycle. It does not come up with a 1000 s tau ADEV by doing that. The gate time would need to be open for 1000 seconds. > > 2) I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading. Clock is used to chunk the data. Is this still true? Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement. Am I understanding this correctly? If you are measuring the difference between two one 1 pps signals, and they are < 0.1 seconds apart, and the TICC has a resolution of 0.1 ns …. Then the reference clock for the TICC only needs to be good to 0.1 ns / 0.1 s = 1x10^-9 > > 3) I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR, Do I even have a chance of doing any meaningful work? (say work with GPSDO and Rb which some of it does reach E-13) Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds? Right now, I don't have any standard that has adev that good at 1 sec anyway. A counter that will do 1x10^-12 at 1 second would need errors below 1 ps. I do not know of any such counter. As you go to other tau’s (like 1,000 seconds) a counter that will keep a gate open for 1000 seconds might do the trick. > > 4) Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have? That may include email and phone call. (I speak English and Japanese) I don't want to lower S/N of this mailing list by doing this here. We have been doing that off list …. > > 5) One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph. How do I tell when a given reading/graph is exceeding the limit of a setup? I did do base line where same signal goes to counter's reference input and signal input. I always have that on my chart so traces does not go below. Is that enough? Look at the counter’s stated accuracy information and do the math. If it’s a 100 ps counter, then that is 100 x 10^-12 at 1 second. If there are no other errors, and the reference is perfect, you will hit 10X better each time the gate increases by 10X. Bob > I appreciate everyone's input. I am learning a lot but just not digesting well enough. I'd like to do DMTD after I understand the basics. > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG > > > On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote: > > I have a question concerning frequency standard and their Allen deviation. (to measure Allen Dev in frequency mode using TimeLab) > > It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else. Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term. > Here's the question part. A frequency counter that measures DUT basically puts out a reading every second during the measurement. When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s. > That being the case, why this consensus of what time source to use for what tau? > I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough. I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME. > I'm still trying to wrap my small head around this. > > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
MD
Magnus Danielson
Sat, Feb 22, 2020 2:25 AM

Hi Taka,

On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote:

I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point?

No problem. No worries. I hope you end up reading these and the other
replies again and acquire good knowledge. I know it's like drinking from
a fire-hoze, but you did ask some very relevant and fair questions.

1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle.

ADEV is about the frequency stability. ADEV can be calculated using
phase or frequency measures. We tend to prefer using phase measures from
Time-Interval Counters for these things.

OK, so let's say that we want to output a counter which provides output
of frequency estimates but for a time-base which is longer than 1 s,
even if we output results every 1 s?

Classically counters could not do that. You acquired a start-value,
waited the time-base, acquired a stop-value, calculated a result to
display and then arm to get a new start-value for the next result. Such
counters will have a limit that the rate of readings will be limited by
the time-base, so if it is set to 10 s, only every 10 s and output is
produced.

To tackle this, one needs a counter that can interleave frequency
measurements, so that it generates new start-points at the update rate
even if the stop-point has not occurred. So, for a time-base of 10 s and
an update rate of 1 s, then every 1 s a new start-trigger is produced,
and then remembered until a stop trigger can be produced, at which time
the start-trigger 10 s back is used to estimate the frequency. In fact,
for this to work, the stop trigger time-stamp is also the start trigger
time-stamp for a new measurement. You can do this with any time-base
really, and the degree of interleaving only depends on the number of
start-points one can keep in memory.

2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly?

There is benefits in time-measures over frequency measures when one
monitors long-term properties. Also, as one tries to create a
phase-curve from frequency estimates, any rounding off errors show in an
slope, as there is a tiny average frequency offset from round-offs. Only
really good such setups does not have significant slope.

3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway.

The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading. The 5370A is
20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12
at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious
improvement. You are more likely to be limited by your oscillators as
ref and under test at 1000 s with that one, than the instrument itself.

4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here. 

I think you have contributed by asking some really good questions.

A setup I use a lot is this:

  1. Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and
    feed into a counter Channel A/TI-start channel. For PPS signals, I make
    sure to trigger a but up on the rising edge not to false-trigger. For
    some counters this means turning of automatic trigger and set it to 1 V
    manually. It is important that no false triggers occurs.

  2. Connect a signal under test to Channel B/TI-stop. Adjust trigger to
    through-zero or up on the edge as suitable.

  3. TI-mode, continuous trigger

  4. Collect data in TimeLab, give TimeLab the frequency of the signal on
    B-channel, give it the time-base of the period on the A-channel.

  5. Look at data as it comes in. Look at phase view, frequency view,
    wrapped phase. Look at the ADEV, how the upper end flaps with data, but
    how the same tau becomes more and more stable as it comes in.

Using even old counters this setup have helped a lot for many measures.
It is simple and sturdy for many measures. Remember to save traces, to
annotate it carefully so one can understand afterwards what one did.

Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and
then PM6853A (2 ns) to show that a particular problem did not needed the
best counter in the house to be well characterized.

5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough?

Almost. It's a good start. The first slope for lower taus is due to the
instrument for sure. If you look carefully you will notice that the
actual performance shifts around, because it is more complex than just
being instrument limit, but it is the right ball-park for that part of
the plots. For the upper end, you can be limited by your device under
test drift. This can be handled by simply letting them be turned on
longer. Sub-sequent measurement will have that rising slope move more
and more towards higher taus and thus becomming less like a limit-issue
for a certain tau.

I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics.

Good spirit. DMTD takes some care, but once you learned it, it can be a
magnificent tool.

Cheers,
Magnus


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

 On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote:  

I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab)

It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term.
Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.
That being the case, why this consensus of what time source to use for what tau?
I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME.
I'm still trying to wrap my small head around this. 


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Taka, On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote: > I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point? No problem. No worries. I hope you end up reading these and the other replies again and acquire good knowledge. I know it's like drinking from a fire-hoze, but you did ask some very relevant and fair questions. > > 1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle. ADEV is about the frequency stability. ADEV can be calculated using phase or frequency measures. We tend to prefer using phase measures from Time-Interval Counters for these things. OK, so let's say that we want to output a counter which provides output of frequency estimates but for a time-base which is longer than 1 s, even if we output results every 1 s? Classically counters could not do that. You acquired a start-value, waited the time-base, acquired a stop-value, calculated a result to display and then arm to get a new start-value for the next result. Such counters will have a limit that the rate of readings will be limited by the time-base, so if it is set to 10 s, only every 10 s and output is produced. To tackle this, one needs a counter that can interleave frequency measurements, so that it generates new start-points at the update rate even if the stop-point has not occurred. So, for a time-base of 10 s and an update rate of 1 s, then every 1 s a new start-trigger is produced, and then remembered until a stop trigger can be produced, at which time the start-trigger 10 s back is used to estimate the frequency. In fact, for this to work, the stop trigger time-stamp is also the start trigger time-stamp for a new measurement. You can do this with any time-base really, and the degree of interleaving only depends on the number of start-points one can keep in memory. > 2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly? There is benefits in time-measures over frequency measures when one monitors long-term properties. Also, as one tries to create a phase-curve from frequency estimates, any rounding off errors show in an slope, as there is a tiny average frequency offset from round-offs. Only really good such setups does not have significant slope. > > 3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway. The resolution of your counter tells you about where your 1/tau curve will cut tau = 1 s, and it goes from there. There is a slight scaling factor, but if we assume it is 1 for now, it is pretty simple. Your 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when the linear slope ends and "lands" in the noise, at which time the noise becomes dominant and is giving you the interesting reading. The 5370A is 20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12 at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious improvement. You are more likely to be limited by your oscillators as ref and under test at 1000 s with that one, than the instrument itself. > 4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here.  I think you have contributed by asking some really good questions. A setup I use a lot is this: 1) Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and feed into a counter Channel A/TI-start channel. For PPS signals, I make sure to trigger a but up on the rising edge not to false-trigger. For some counters this means turning of automatic trigger and set it to 1 V manually. It is important that no false triggers occurs. 2) Connect a signal under test to Channel B/TI-stop. Adjust trigger to through-zero or up on the edge as suitable. 3) TI-mode, continuous trigger 4) Collect data in TimeLab, give TimeLab the frequency of the signal on B-channel, give it the time-base of the period on the A-channel. 5) Look at data as it comes in. Look at phase view, frequency view, wrapped phase. Look at the ADEV, how the upper end flaps with data, but how the same tau becomes more and more stable as it comes in. Using even old counters this setup have helped a lot for many measures. It is simple and sturdy for many measures. Remember to save traces, to annotate it carefully so one can understand afterwards what one did. Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and then PM6853A (2 ns) to show that a particular problem did not needed the best counter in the house to be well characterized. > > 5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough? Almost. It's a good start. The first slope for lower taus is due to the instrument for sure. If you look carefully you will notice that the actual performance shifts around, because it is more complex than just being instrument limit, but it is the right ball-park for that part of the plots. For the upper end, you can be limited by your device under test drift. This can be handled by simply letting them be turned on longer. Sub-sequent measurement will have that rising slope move more and more towards higher taus and thus becomming less like a limit-issue for a certain tau. > I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics. Good spirit. DMTD takes some care, but once you learned it, it can be a magnificent tool. Cheers, Magnus > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG > > > On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote: > > I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab) > > It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term. > Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s. > That being the case, why this consensus of what time source to use for what tau? > I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME. > I'm still trying to wrap my small head around this.  > > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
TK
Taka Kamiya
Sat, Feb 22, 2020 3:55 AM

It's not like drinking from a fire hydrant.  It's like drowning in hoover dam, get sucked into an inlet, pulverized by turbine blade, and getting spit out into a stream. 

One question : 

You said this:
"The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading."
By using the same logic, I can keep going up and up on longer gate time and tau keeps getting better and better.  I know at one point, inflection happens and that indicates noise taking over.  But what kind of noise (phase?) and how does that happen?  Only thing that changes in this equation is the gate time.  Everything is constant.  You mean gate time is no longer accurate enough to support the minute shift in phase?
I'm still confused about the precision (not accuracy) of the time base.  Am I still ultimately constrained by this?  Without DMTD, or some kind of pre-scaling of DUT, if I measure Rb with time base using another Rb, they are both rubber-bands, correct?
I'm infinitely curious by nature.  I need to know everything, even to a minute detail, to be satisfied.  I hope you don't get tired of this.   

(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

On Friday, February 21, 2020, 9:26:47 PM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote:  

Hi Taka,

On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote:

I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point?

No problem. No worries. I hope you end up reading these and the other
replies again and acquire good knowledge. I know it's like drinking from
a fire-hoze, but you did ask some very relevant and fair questions.

1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle.

ADEV is about the frequency stability. ADEV can be calculated using
phase or frequency measures. We tend to prefer using phase measures from
Time-Interval Counters for these things.

OK, so let's say that we want to output a counter which provides output
of frequency estimates but for a time-base which is longer than 1 s,
even if we output results every 1 s?

Classically counters could not do that. You acquired a start-value,
waited the time-base, acquired a stop-value, calculated a result to
display and then arm to get a new start-value for the next result. Such
counters will have a limit that the rate of readings will be limited by
the time-base, so if it is set to 10 s, only every 10 s and output is
produced.

To tackle this, one needs a counter that can interleave frequency
measurements, so that it generates new start-points at the update rate
even if the stop-point has not occurred. So, for a time-base of 10 s and
an update rate of 1 s, then every 1 s a new start-trigger is produced,
and then remembered until a stop trigger can be produced, at which time
the start-trigger 10 s back is used to estimate the frequency. In fact,
for this to work, the stop trigger time-stamp is also the start trigger
time-stamp for a new measurement. You can do this with any time-base
really, and the degree of interleaving only depends on the number of
start-points one can keep in memory.

2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly?

There is benefits in time-measures over frequency measures when one
monitors long-term properties. Also, as one tries to create a
phase-curve from frequency estimates, any rounding off errors show in an
slope, as there is a tiny average frequency offset from round-offs. Only
really good such setups does not have significant slope.

3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway.

The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading. The 5370A is
20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12
at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious
improvement. You are more likely to be limited by your oscillators as
ref and under test at 1000 s with that one, than the instrument itself.

4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here. 

I think you have contributed by asking some really good questions.

A setup I use a lot is this:

  1. Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and
    feed into a counter Channel A/TI-start channel. For PPS signals, I make
    sure to trigger a but up on the rising edge not to false-trigger. For
    some counters this means turning of automatic trigger and set it to 1 V
    manually. It is important that no false triggers occurs.

  2. Connect a signal under test to Channel B/TI-stop. Adjust trigger to
    through-zero or up on the edge as suitable.

  3. TI-mode, continuous trigger

  4. Collect data in TimeLab, give TimeLab the frequency of the signal on
    B-channel, give it the time-base of the period on the A-channel.

  5. Look at data as it comes in. Look at phase view, frequency view,
    wrapped phase. Look at the ADEV, how the upper end flaps with data, but
    how the same tau becomes more and more stable as it comes in.

Using even old counters this setup have helped a lot for many measures.
It is simple and sturdy for many measures. Remember to save traces, to
annotate it carefully so one can understand afterwards what one did.

Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and
then PM6853A (2 ns) to show that a particular problem did not needed the
best counter in the house to be well characterized.

5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough?

Almost. It's a good start. The first slope for lower taus is due to the
instrument for sure. If you look carefully you will notice that the
actual performance shifts around, because it is more complex than just
being instrument limit, but it is the right ball-park for that part of
the plots. For the upper end, you can be limited by your device under
test drift. This can be handled by simply letting them be turned on
longer. Sub-sequent measurement will have that rising slope move more
and more towards higher taus and thus becomming less like a limit-issue
for a certain tau.

I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics.

Good spirit. DMTD takes some care, but once you learned it, it can be a
magnificent tool.

Cheers,
Magnus


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
 

    On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts time-nuts@lists.febo.com wrote: 
 
  I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab)

It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term.
Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.
That being the case, why this consensus of what time source to use for what tau?
I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME.
I'm still trying to wrap my small head around this. 


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
   


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

It's not like drinking from a fire hydrant.  It's like drowning in hoover dam, get sucked into an inlet, pulverized by turbine blade, and getting spit out into a stream.  One question :  You said this: "The resolution of your counter tells you about where your 1/tau curve will cut tau = 1 s, and it goes from there. There is a slight scaling factor, but if we assume it is 1 for now, it is pretty simple. Your 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when the linear slope ends and "lands" in the noise, at which time the noise becomes dominant and is giving you the interesting reading." By using the same logic, I can keep going up and up on longer gate time and tau keeps getting better and better.  I know at one point, inflection happens and that indicates noise taking over.  But what kind of noise (phase?) and how does that happen?  Only thing that changes in this equation is the gate time.  Everything is constant.  You mean gate time is no longer accurate enough to support the minute shift in phase? I'm still confused about the precision (not accuracy) of the time base.  Am I still ultimately constrained by this?  Without DMTD, or some kind of pre-scaling of DUT, if I measure Rb with time base using another Rb, they are both rubber-bands, correct? I'm infinitely curious by nature.  I need to know everything, even to a minute detail, to be satisfied.  I hope you don't get tired of this.    --------------------------------------- (Mr.) Taka Kamiya KB4EMF / ex JF2DKG On Friday, February 21, 2020, 9:26:47 PM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote: Hi Taka, On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote: > I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point? No problem. No worries. I hope you end up reading these and the other replies again and acquire good knowledge. I know it's like drinking from a fire-hoze, but you did ask some very relevant and fair questions. > > 1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle. ADEV is about the frequency stability. ADEV can be calculated using phase or frequency measures. We tend to prefer using phase measures from Time-Interval Counters for these things. OK, so let's say that we want to output a counter which provides output of frequency estimates but for a time-base which is longer than 1 s, even if we output results every 1 s? Classically counters could not do that. You acquired a start-value, waited the time-base, acquired a stop-value, calculated a result to display and then arm to get a new start-value for the next result. Such counters will have a limit that the rate of readings will be limited by the time-base, so if it is set to 10 s, only every 10 s and output is produced. To tackle this, one needs a counter that can interleave frequency measurements, so that it generates new start-points at the update rate even if the stop-point has not occurred. So, for a time-base of 10 s and an update rate of 1 s, then every 1 s a new start-trigger is produced, and then remembered until a stop trigger can be produced, at which time the start-trigger 10 s back is used to estimate the frequency. In fact, for this to work, the stop trigger time-stamp is also the start trigger time-stamp for a new measurement. You can do this with any time-base really, and the degree of interleaving only depends on the number of start-points one can keep in memory. > 2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly? There is benefits in time-measures over frequency measures when one monitors long-term properties. Also, as one tries to create a phase-curve from frequency estimates, any rounding off errors show in an slope, as there is a tiny average frequency offset from round-offs. Only really good such setups does not have significant slope. > > 3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway. The resolution of your counter tells you about where your 1/tau curve will cut tau = 1 s, and it goes from there. There is a slight scaling factor, but if we assume it is 1 for now, it is pretty simple. Your 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when the linear slope ends and "lands" in the noise, at which time the noise becomes dominant and is giving you the interesting reading. The 5370A is 20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12 at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious improvement. You are more likely to be limited by your oscillators as ref and under test at 1000 s with that one, than the instrument itself. > 4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here.  I think you have contributed by asking some really good questions. A setup I use a lot is this: 1) Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and feed into a counter Channel A/TI-start channel. For PPS signals, I make sure to trigger a but up on the rising edge not to false-trigger. For some counters this means turning of automatic trigger and set it to 1 V manually. It is important that no false triggers occurs. 2) Connect a signal under test to Channel B/TI-stop. Adjust trigger to through-zero or up on the edge as suitable. 3) TI-mode, continuous trigger 4) Collect data in TimeLab, give TimeLab the frequency of the signal on B-channel, give it the time-base of the period on the A-channel. 5) Look at data as it comes in. Look at phase view, frequency view, wrapped phase. Look at the ADEV, how the upper end flaps with data, but how the same tau becomes more and more stable as it comes in. Using even old counters this setup have helped a lot for many measures. It is simple and sturdy for many measures. Remember to save traces, to annotate it carefully so one can understand afterwards what one did. Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and then PM6853A (2 ns) to show that a particular problem did not needed the best counter in the house to be well characterized. > > 5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough? Almost. It's a good start. The first slope for lower taus is due to the instrument for sure. If you look carefully you will notice that the actual performance shifts around, because it is more complex than just being instrument limit, but it is the right ball-park for that part of the plots. For the upper end, you can be limited by your device under test drift. This can be handled by simply letting them be turned on longer. Sub-sequent measurement will have that rising slope move more and more towards higher taus and thus becomming less like a limit-issue for a certain tau. > I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics. Good spirit. DMTD takes some care, but once you learned it, it can be a magnificent tool. Cheers, Magnus > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG >  > >    On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote:  >  >  I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab) > > It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term. > Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s. > That being the case, why this consensus of what time source to use for what tau? > I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME. > I'm still trying to wrap my small head around this.  > > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >    > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow the instructions there.
MD
Magnus Danielson
Sat, Feb 22, 2020 8:02 AM

Hi Taka,

On 2020-02-22 04:55, Taka Kamiya via time-nuts wrote:

It's not like drinking from a fire hydrant.  It's like drowning in hoover dam, get sucked into an inlet, pulverized by turbine blade, and getting spit out into a stream. 

Hmm, not exactly what I had hoped for as your experience... whatever,
hope it is worth it.

One question : 

You said this:
"The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading."
By using the same logic, I can keep going up and up on longer gate time and tau keeps getting better and better.  I know at one point, inflection happens and that indicates noise taking over.  But what kind of noise (phase?) and how does that happen?

OK. Good question.

The instrument noise in itself can be while phase noise and flicker
phase noise. The resolution limit will be a systematic disturbance. All
these three has the same 1/tau slope so from these three, nothing
happens as tau increases. At some point naturally will thermal
instability come in, but that will be another systematic, but it will be
hard to cancel. So that is the limit of the instrument.

However, the noise of the reference and DUT will dominate for larger
taus, and thus it will be good enough to measure that noise. Besides,
length of measurement becomes a limit.

  Only thing that changes in this equation is the gate time.  Everything is constant.  You mean gate time is no longer accurate enough to support the minute shift in phase?
I'm still confused about the precision (not accuracy) of the time base.  Am I still ultimately constrained by this?

There are ways to get around it.

  Without DMTD, or some kind of pre-scaling of DUT, if I measure Rb with time base using another Rb, they are both rubber-bands, correct?

Now, what you can do is to do a three-cornered hat measurement.
Essentially you measure three sources at once as three pair-wise
measurements. You calculate your ADEV of each pair. Now, as you have the
noises n1, n2 and n3 of each source, measurement m1 = n1+n2, m2 = n1+n3
and m3 = n2+n3. This equation problem can be solved to bring out n1, n2
and n3 separate. The trouble is, this does not always work out as
precisely as one would like, because the measurements isn't precise and
well converged. Therefore it is a bit of sensitive process, but if one
gets it working, you can achieve the separation.

I'm infinitely curious by nature.  I need to know everything, even to a minute detail, to be satisfied.  I hope you don't get tired of this.   

So far, you ask relevant questions that can be given an answer. Being a
curious person myself, I've digged down.

Cheers,
Magnus

Hi Taka, On 2020-02-22 04:55, Taka Kamiya via time-nuts wrote: > It's not like drinking from a fire hydrant.  It's like drowning in hoover dam, get sucked into an inlet, pulverized by turbine blade, and getting spit out into a stream.  Hmm, not exactly what I had hoped for as your experience... whatever, hope it is worth it. > > One question :  > > You said this: > "The resolution of your counter tells you about where your 1/tau curve > will cut tau = 1 s, and it goes from there. There is a slight scaling > factor, but if we assume it is 1 for now, it is pretty simple. Your > 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 > at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when > the linear slope ends and "lands" in the noise, at which time the noise > becomes dominant and is giving you the interesting reading." > By using the same logic, I can keep going up and up on longer gate time and tau keeps getting better and better.  I know at one point, inflection happens and that indicates noise taking over.  But what kind of noise (phase?) and how does that happen? OK. Good question. The instrument noise in itself can be while phase noise and flicker phase noise. The resolution limit will be a systematic disturbance. All these three has the same 1/tau slope so from these three, nothing happens as tau increases. At some point naturally will thermal instability come in, but that will be another systematic, but it will be hard to cancel. So that is the limit of the instrument. However, the noise of the reference and DUT will dominate for larger taus, and thus it will be good enough to measure that noise. Besides, length of measurement becomes a limit. >   Only thing that changes in this equation is the gate time.  Everything is constant.  You mean gate time is no longer accurate enough to support the minute shift in phase? > I'm still confused about the precision (not accuracy) of the time base.  Am I still ultimately constrained by this? There are ways to get around it. >   Without DMTD, or some kind of pre-scaling of DUT, if I measure Rb with time base using another Rb, they are both rubber-bands, correct? Now, what you can do is to do a three-cornered hat measurement. Essentially you measure three sources at once as three pair-wise measurements. You calculate your ADEV of each pair. Now, as you have the noises n1, n2 and n3 of each source, measurement m1 = n1+n2, m2 = n1+n3 and m3 = n2+n3. This equation problem can be solved to bring out n1, n2 and n3 separate. The trouble is, this does not always work out as precisely as one would like, because the measurements isn't precise and well converged. Therefore it is a bit of sensitive process, but if one gets it working, you can achieve the separation. > I'm infinitely curious by nature.  I need to know everything, even to a minute detail, to be satisfied.  I hope you don't get tired of this.    So far, you ask relevant questions that can be given an answer. Being a curious person myself, I've digged down. Cheers, Magnus
TK
Taka Kamiya
Sat, Feb 22, 2020 3:08 PM

As I understand, 3 cornered hat is nothing more than system of equations where you have 3 variables and 3 equations, that allows for solving for each.  I'm going to hold off on that one.  I forgot about thermal and equipment's internal noise.  It all makes sense.  I hope to be doing some measuring this weekend if all works as planned (hope, hope, hope....)
I mentioned earlier, I have various standards.  I also have TAPR's board to create 1 pps from any standards I have.  To start and follow your last example on how-to, what would you recommend for DUT and the standard?  I'd like to do one that I can show you, and immediately say if it's what you'd expect or not.
I have newly repaired HP105B, PRS10, Efratom FRK-c?, LPRO-101, and various GPSDO including T-bolt proper.  For the counter, I think I'll use the old HP interval counter, rather than 50132A.  The latter is all menu driven.  It's so easy to mis-set or otherwise screw it up and not notice it.  HP5370A is all switches and knobs.  How long of measurement?  In my experience, 2 days will show the inflection point. 


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

On Saturday, February 22, 2020, 3:03:37 AM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote:  

Hi Taka,

On 2020-02-22 04:55, Taka Kamiya via time-nuts wrote:

It's not like drinking from a fire hydrant.  It's like drowning in hoover dam, get sucked into an inlet, pulverized by turbine blade, and getting spit out into a stream. 

Hmm, not exactly what I had hoped for as your experience... whatever,
hope it is worth it.

One question : 

You said this:
"The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading."
By using the same logic, I can keep going up and up on longer gate time and tau keeps getting better and better.  I know at one point, inflection happens and that indicates noise taking over.  But what kind of noise (phase?) and how does that happen?

OK. Good question.

The instrument noise in itself can be while phase noise and flicker
phase noise. The resolution limit will be a systematic disturbance. All
these three has the same 1/tau slope so from these three, nothing
happens as tau increases. At some point naturally will thermal
instability come in, but that will be another systematic, but it will be
hard to cancel. So that is the limit of the instrument.

However, the noise of the reference and DUT will dominate for larger
taus, and thus it will be good enough to measure that noise. Besides,
length of measurement becomes a limit.

  Only thing that changes in this equation is the gate time.  Everything is constant.  You mean gate time is no longer accurate enough to support the minute shift in phase?
I'm still confused about the precision (not accuracy) of the time base.  Am I still ultimately constrained by this?

There are ways to get around it.

  Without DMTD, or some kind of pre-scaling of DUT, if I measure Rb with time base using another Rb, they are both rubber-bands, correct?

Now, what you can do is to do a three-cornered hat measurement.
Essentially you measure three sources at once as three pair-wise
measurements. You calculate your ADEV of each pair. Now, as you have the
noises n1, n2 and n3 of each source, measurement m1 = n1+n2, m2 = n1+n3
and m3 = n2+n3. This equation problem can be solved to bring out n1, n2
and n3 separate. The trouble is, this does not always work out as
precisely as one would like, because the measurements isn't precise and
well converged. Therefore it is a bit of sensitive process, but if one
gets it working, you can achieve the separation.

I'm infinitely curious by nature.  I need to know everything, even to a minute detail, to be satisfied.  I hope you don't get tired of this.   

So far, you ask relevant questions that can be given an answer. Being a
curious person myself, I've digged down.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

As I understand, 3 cornered hat is nothing more than system of equations where you have 3 variables and 3 equations, that allows for solving for each.  I'm going to hold off on that one.  I forgot about thermal and equipment's internal noise.  It all makes sense.  I hope to be doing some measuring this weekend if all works as planned (hope, hope, hope....) I mentioned earlier, I have various standards.  I also have TAPR's board to create 1 pps from any standards I have.  To start and follow your last example on how-to, what would you recommend for DUT and the standard?  I'd like to do one that I can show you, and immediately say if it's what you'd expect or not. I have newly repaired HP105B, PRS10, Efratom FRK-c?, LPRO-101, and various GPSDO including T-bolt proper.  For the counter, I think I'll use the old HP interval counter, rather than 50132A.  The latter is all menu driven.  It's so easy to mis-set or otherwise screw it up and not notice it.  HP5370A is all switches and knobs.  How long of measurement?  In my experience, 2 days will show the inflection point.  --------------------------------------- (Mr.) Taka Kamiya KB4EMF / ex JF2DKG On Saturday, February 22, 2020, 3:03:37 AM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote: Hi Taka, On 2020-02-22 04:55, Taka Kamiya via time-nuts wrote: > It's not like drinking from a fire hydrant.  It's like drowning in hoover dam, get sucked into an inlet, pulverized by turbine blade, and getting spit out into a stream.  Hmm, not exactly what I had hoped for as your experience... whatever, hope it is worth it. > > One question :  > > You said this: > "The resolution of your counter tells you about where your 1/tau curve > will cut tau = 1 s, and it goes from there. There is a slight scaling > factor, but if we assume it is 1 for now, it is pretty simple. Your > 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 > at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when > the linear slope ends and "lands" in the noise, at which time the noise > becomes dominant and is giving you the interesting reading." > By using the same logic, I can keep going up and up on longer gate time and tau keeps getting better and better.  I know at one point, inflection happens and that indicates noise taking over.  But what kind of noise (phase?) and how does that happen? OK. Good question. The instrument noise in itself can be while phase noise and flicker phase noise. The resolution limit will be a systematic disturbance. All these three has the same 1/tau slope so from these three, nothing happens as tau increases. At some point naturally will thermal instability come in, but that will be another systematic, but it will be hard to cancel. So that is the limit of the instrument. However, the noise of the reference and DUT will dominate for larger taus, and thus it will be good enough to measure that noise. Besides, length of measurement becomes a limit. >   Only thing that changes in this equation is the gate time.  Everything is constant.  You mean gate time is no longer accurate enough to support the minute shift in phase? > I'm still confused about the precision (not accuracy) of the time base.  Am I still ultimately constrained by this? There are ways to get around it. >   Without DMTD, or some kind of pre-scaling of DUT, if I measure Rb with time base using another Rb, they are both rubber-bands, correct? Now, what you can do is to do a three-cornered hat measurement. Essentially you measure three sources at once as three pair-wise measurements. You calculate your ADEV of each pair. Now, as you have the noises n1, n2 and n3 of each source, measurement m1 = n1+n2, m2 = n1+n3 and m3 = n2+n3. This equation problem can be solved to bring out n1, n2 and n3 separate. The trouble is, this does not always work out as precisely as one would like, because the measurements isn't precise and well converged. Therefore it is a bit of sensitive process, but if one gets it working, you can achieve the separation. > I'm infinitely curious by nature.  I need to know everything, even to a minute detail, to be satisfied.  I hope you don't get tired of this.    So far, you ask relevant questions that can be given an answer. Being a curious person myself, I've digged down. Cheers, Magnus _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow the instructions there.
MD
Magnus Danielson
Sat, Feb 22, 2020 3:51 PM

Hi Taka,

On 2020-02-22 16:08, Taka Kamiya via time-nuts wrote:

As I understand, 3 cornered hat is nothing more than system of equations where you have 3 variables and 3 equations, that allows for solving for each.  I'm going to hold off on that one.  I forgot about thermal and equipment's internal noise.  It all makes sense.  I hope to be doing some measuring this weekend if all works as planned (hope, hope, hope....)

Good. As you get measures, more and more of this will sink in as you see
them.

I mentioned earlier, I have various standards.  I also have TAPR's board to create 1 pps from any standards I have.  To start and follow your last example on how-to, what would you recommend for DUT and the standard?  I'd like to do one that I can show you, and immediately say if it's what you'd expect or not.
I have newly repaired HP105B, PRS10, Efratom FRK-c?, LPRO-101, and various GPSDO including T-bolt proper.  For the counter, I think I'll use the old HP interval counter, rather than 50132A.  The latter is all menu driven.  It's so easy to mis-set or otherwise screw it up and not notice it.  HP5370A is all switches and knobs.  How long of measurement?  In my experience, 2 days will show the inflection point. 

I would start doing a few 1 hour measurements. Only when the measurement
makes sense you extend it.

The HP53132A requires you to set trigger to manual and then adjust
trigger level manually. A little painful in the beginning, but you learn
to do it. In the beginning it can be good to use an oscilloscope to
learn where a good trigger point is.

HP5370 will work. Use PPS from one of your GPSDOs for start. Use any
oscillator of choice as DUT. Once the measurement on one makes sense,
keep that and go to the next.

Mistakes in setup causes huge face-jumps, which is why wrapped phase
(w), phase (p) and frequency (f) is useful as one learns. After learning
to set things up, one end up not using wrapped phase anymore, one sees
the error and just fixes it. Use the 10 MHz output of a counter with no
external clock attached as a clock with worse perfromance than rubidiums.

Cheers,
Magnus


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

 On Saturday, February 22, 2020, 3:03:37 AM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote:  

Hi Taka,

On 2020-02-22 04:55, Taka Kamiya via time-nuts wrote:

It's not like drinking from a fire hydrant.  It's like drowning in hoover dam, get sucked into an inlet, pulverized by turbine blade, and getting spit out into a stream. 

Hmm, not exactly what I had hoped for as your experience... whatever,
hope it is worth it.

One question : 

You said this:
"The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading."
By using the same logic, I can keep going up and up on longer gate time and tau keeps getting better and better.  I know at one point, inflection happens and that indicates noise taking over.  But what kind of noise (phase?) and how does that happen?

OK. Good question.

The instrument noise in itself can be while phase noise and flicker
phase noise. The resolution limit will be a systematic disturbance. All
these three has the same 1/tau slope so from these three, nothing
happens as tau increases. At some point naturally will thermal
instability come in, but that will be another systematic, but it will be
hard to cancel. So that is the limit of the instrument.

However, the noise of the reference and DUT will dominate for larger
taus, and thus it will be good enough to measure that noise. Besides,
length of measurement becomes a limit.

  Only thing that changes in this equation is the gate time.  Everything is constant.  You mean gate time is no longer accurate enough to support the minute shift in phase?
I'm still confused about the precision (not accuracy) of the time base.  Am I still ultimately constrained by this?

There are ways to get around it.

  Without DMTD, or some kind of pre-scaling of DUT, if I measure Rb with time base using another Rb, they are both rubber-bands, correct?

Now, what you can do is to do a three-cornered hat measurement.
Essentially you measure three sources at once as three pair-wise
measurements. You calculate your ADEV of each pair. Now, as you have the
noises n1, n2 and n3 of each source, measurement m1 = n1+n2, m2 = n1+n3
and m3 = n2+n3. This equation problem can be solved to bring out n1, n2
and n3 separate. The trouble is, this does not always work out as
precisely as one would like, because the measurements isn't precise and
well converged. Therefore it is a bit of sensitive process, but if one
gets it working, you can achieve the separation.

I'm infinitely curious by nature.  I need to know everything, even to a minute detail, to be satisfied.  I hope you don't get tired of this.   

So far, you ask relevant questions that can be given an answer. Being a
curious person myself, I've digged down.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Taka, On 2020-02-22 16:08, Taka Kamiya via time-nuts wrote: > As I understand, 3 cornered hat is nothing more than system of equations where you have 3 variables and 3 equations, that allows for solving for each.  I'm going to hold off on that one.  I forgot about thermal and equipment's internal noise.  It all makes sense.  I hope to be doing some measuring this weekend if all works as planned (hope, hope, hope....) Good. As you get measures, more and more of this will sink in as you see them. > I mentioned earlier, I have various standards.  I also have TAPR's board to create 1 pps from any standards I have.  To start and follow your last example on how-to, what would you recommend for DUT and the standard?  I'd like to do one that I can show you, and immediately say if it's what you'd expect or not. > I have newly repaired HP105B, PRS10, Efratom FRK-c?, LPRO-101, and various GPSDO including T-bolt proper.  For the counter, I think I'll use the old HP interval counter, rather than 50132A.  The latter is all menu driven.  It's so easy to mis-set or otherwise screw it up and not notice it.  HP5370A is all switches and knobs.  How long of measurement?  In my experience, 2 days will show the inflection point.  I would start doing a few 1 hour measurements. Only when the measurement makes sense you extend it. The HP53132A requires you to set trigger to manual and then adjust trigger level manually. A little painful in the beginning, but you learn to do it. In the beginning it can be good to use an oscilloscope to learn where a good trigger point is. HP5370 will work. Use PPS from one of your GPSDOs for start. Use any oscillator of choice as DUT. Once the measurement on one makes sense, keep that and go to the next. Mistakes in setup causes huge face-jumps, which is why wrapped phase (w), phase (p) and frequency (f) is useful as one learns. After learning to set things up, one end up not using wrapped phase anymore, one sees the error and just fixes it. Use the 10 MHz output of a counter with no external clock attached as a clock with worse perfromance than rubidiums. Cheers, Magnus > > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG > > > On Saturday, February 22, 2020, 3:03:37 AM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote: > > Hi Taka, > > On 2020-02-22 04:55, Taka Kamiya via time-nuts wrote: >> It's not like drinking from a fire hydrant.  It's like drowning in hoover dam, get sucked into an inlet, pulverized by turbine blade, and getting spit out into a stream.  > Hmm, not exactly what I had hoped for as your experience... whatever, > hope it is worth it. >> One question :  >> >> You said this: >> "The resolution of your counter tells you about where your 1/tau curve >> will cut tau = 1 s, and it goes from there. There is a slight scaling >> factor, but if we assume it is 1 for now, it is pretty simple. Your >> 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 >> at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when >> the linear slope ends and "lands" in the noise, at which time the noise >> becomes dominant and is giving you the interesting reading." >> By using the same logic, I can keep going up and up on longer gate time and tau keeps getting better and better.  I know at one point, inflection happens and that indicates noise taking over.  But what kind of noise (phase?) and how does that happen? > OK. Good question. > > The instrument noise in itself can be while phase noise and flicker > phase noise. The resolution limit will be a systematic disturbance. All > these three has the same 1/tau slope so from these three, nothing > happens as tau increases. At some point naturally will thermal > instability come in, but that will be another systematic, but it will be > hard to cancel. So that is the limit of the instrument. > > However, the noise of the reference and DUT will dominate for larger > taus, and thus it will be good enough to measure that noise. Besides, > length of measurement becomes a limit. > >>   Only thing that changes in this equation is the gate time.  Everything is constant.  You mean gate time is no longer accurate enough to support the minute shift in phase? >> I'm still confused about the precision (not accuracy) of the time base.  Am I still ultimately constrained by this? > There are ways to get around it. >>   Without DMTD, or some kind of pre-scaling of DUT, if I measure Rb with time base using another Rb, they are both rubber-bands, correct? > Now, what you can do is to do a three-cornered hat measurement. > Essentially you measure three sources at once as three pair-wise > measurements. You calculate your ADEV of each pair. Now, as you have the > noises n1, n2 and n3 of each source, measurement m1 = n1+n2, m2 = n1+n3 > and m3 = n2+n3. This equation problem can be solved to bring out n1, n2 > and n3 separate. The trouble is, this does not always work out as > precisely as one would like, because the measurements isn't precise and > well converged. Therefore it is a bit of sensitive process, but if one > gets it working, you can achieve the separation. >> I'm infinitely curious by nature.  I need to know everything, even to a minute detail, to be satisfied.  I hope you don't get tired of this.    > So far, you ask relevant questions that can be given an answer. Being a > curious person myself, I've digged down. > > Cheers, > Magnus > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
TK
Taka Kamiya
Sat, Feb 29, 2020 10:10 PM

One question for Magnus.

Ch A start - pps (standard)Ch B stop - DUT
On item 4, you said "frequency of the signal on time B".  That much is obvious.  But then you said: "give it the time-base of the period on the A-channel".  Will you explain this?
Say I give 1 Hz, period is 1s.  Say I give 10Hz, period is 0.1s.  Is this what you mean?
I'm using HP5370A.  This instruction is valid on this TI counter, correct??

A request for everyone:
I am conducting an one hour measurement on HP105B.  Does anyone have 1 hour plot of this signal generator handy?  If so, will you DM me a copy?  For some reason, I cannot find one on the great Internet.

------------------- clip from Magnus's previous email------------------
A setup I use a lot is this:

  1. Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and
    feed into a counter Channel A/TI-start channel. For PPS signals, I make
    sure to trigger a but up on the rising edge not to false-trigger. For
    some counters this means turning of automatic trigger and set it to 1 V
    manually. It is important that no false triggers occurs.

  2. Connect a signal under test to Channel B/TI-stop. Adjust trigger to
    through-zero or up on the edge as suitable.

  3. TI-mode, continuous trigger

  4. Collect data in TimeLab, give TimeLab the frequency of the signal on
    B-channel, give it the time-base of the period on the A-channel.

  5. Look at data as it comes in. Look at phase view, frequency view,
    wrapped phase. Look at the ADEV, how the upper end flaps with data, but
    how the same tau becomes more and more stable as it comes in.---------------------


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

On Friday, February 21, 2020, 9:26:47 PM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote:  

Hi Taka,

On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote:

I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point?

No problem. No worries. I hope you end up reading these and the other
replies again and acquire good knowledge. I know it's like drinking from
a fire-hoze, but you did ask some very relevant and fair questions.

1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle.

ADEV is about the frequency stability. ADEV can be calculated using
phase or frequency measures. We tend to prefer using phase measures from
Time-Interval Counters for these things.

OK, so let's say that we want to output a counter which provides output
of frequency estimates but for a time-base which is longer than 1 s,
even if we output results every 1 s?

Classically counters could not do that. You acquired a start-value,
waited the time-base, acquired a stop-value, calculated a result to
display and then arm to get a new start-value for the next result. Such
counters will have a limit that the rate of readings will be limited by
the time-base, so if it is set to 10 s, only every 10 s and output is
produced.

To tackle this, one needs a counter that can interleave frequency
measurements, so that it generates new start-points at the update rate
even if the stop-point has not occurred. So, for a time-base of 10 s and
an update rate of 1 s, then every 1 s a new start-trigger is produced,
and then remembered until a stop trigger can be produced, at which time
the start-trigger 10 s back is used to estimate the frequency. In fact,
for this to work, the stop trigger time-stamp is also the start trigger
time-stamp for a new measurement. You can do this with any time-base
really, and the degree of interleaving only depends on the number of
start-points one can keep in memory.

2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly?

There is benefits in time-measures over frequency measures when one
monitors long-term properties. Also, as one tries to create a
phase-curve from frequency estimates, any rounding off errors show in an
slope, as there is a tiny average frequency offset from round-offs. Only
really good such setups does not have significant slope.

3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway.

The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading. The 5370A is
20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12
at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious
improvement. You are more likely to be limited by your oscillators as
ref and under test at 1000 s with that one, than the instrument itself.

4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here. 

I think you have contributed by asking some really good questions.

A setup I use a lot is this:

  1. Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and
    feed into a counter Channel A/TI-start channel. For PPS signals, I make
    sure to trigger a but up on the rising edge not to false-trigger. For
    some counters this means turning of automatic trigger and set it to 1 V
    manually. It is important that no false triggers occurs.

  2. Connect a signal under test to Channel B/TI-stop. Adjust trigger to
    through-zero or up on the edge as suitable.

  3. TI-mode, continuous trigger

  4. Collect data in TimeLab, give TimeLab the frequency of the signal on
    B-channel, give it the time-base of the period on the A-channel.

  5. Look at data as it comes in. Look at phase view, frequency view,
    wrapped phase. Look at the ADEV, how the upper end flaps with data, but
    how the same tau becomes more and more stable as it comes in.

Using even old counters this setup have helped a lot for many measures.
It is simple and sturdy for many measures. Remember to save traces, to
annotate it carefully so one can understand afterwards what one did.

Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and
then PM6853A (2 ns) to show that a particular problem did not needed the
best counter in the house to be well characterized.

5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough?

Almost. It's a good start. The first slope for lower taus is due to the
instrument for sure. If you look carefully you will notice that the
actual performance shifts around, because it is more complex than just
being instrument limit, but it is the right ball-park for that part of
the plots. For the upper end, you can be limited by your device under
test drift. This can be handled by simply letting them be turned on
longer. Sub-sequent measurement will have that rising slope move more
and more towards higher taus and thus becomming less like a limit-issue
for a certain tau.

I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics.

Good spirit. DMTD takes some care, but once you learned it, it can be a
magnificent tool.

Cheers,
Magnus


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
 

    On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts time-nuts@lists.febo.com wrote: 
 
  I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab)

It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term.
Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.
That being the case, why this consensus of what time source to use for what tau?
I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME.
I'm still trying to wrap my small head around this. 


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
   


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

One question for Magnus. Ch A start - pps (standard)Ch B stop - DUT On item 4, you said "frequency of the signal on time B".  That much is obvious.  But then you said: "give it the time-base of the period on the A-channel".  Will you explain this? Say I give 1 Hz, period is 1s.  Say I give 10Hz, period is 0.1s.  Is this what you mean? I'm using HP5370A.  This instruction is valid on this TI counter, correct?? A request for everyone: I am conducting an one hour measurement on HP105B.  Does anyone have 1 hour plot of this signal generator handy?  If so, will you DM me a copy?  For some reason, I cannot find one on the great Internet. ------------------- clip from Magnus's previous email------------------ A setup I use a lot is this: 1) Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and feed into a counter Channel A/TI-start channel. For PPS signals, I make sure to trigger a but up on the rising edge not to false-trigger. For some counters this means turning of automatic trigger and set it to 1 V manually. It is important that no false triggers occurs. 2) Connect a signal under test to Channel B/TI-stop. Adjust trigger to through-zero or up on the edge as suitable. 3) TI-mode, continuous trigger 4) Collect data in TimeLab, give TimeLab the frequency of the signal on B-channel, give it the time-base of the period on the A-channel. 5) Look at data as it comes in. Look at phase view, frequency view, wrapped phase. Look at the ADEV, how the upper end flaps with data, but how the same tau becomes more and more stable as it comes in.--------------------- --------------------------------------- (Mr.) Taka Kamiya KB4EMF / ex JF2DKG On Friday, February 21, 2020, 9:26:47 PM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote: Hi Taka, On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote: > I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point? No problem. No worries. I hope you end up reading these and the other replies again and acquire good knowledge. I know it's like drinking from a fire-hoze, but you did ask some very relevant and fair questions. > > 1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle. ADEV is about the frequency stability. ADEV can be calculated using phase or frequency measures. We tend to prefer using phase measures from Time-Interval Counters for these things. OK, so let's say that we want to output a counter which provides output of frequency estimates but for a time-base which is longer than 1 s, even if we output results every 1 s? Classically counters could not do that. You acquired a start-value, waited the time-base, acquired a stop-value, calculated a result to display and then arm to get a new start-value for the next result. Such counters will have a limit that the rate of readings will be limited by the time-base, so if it is set to 10 s, only every 10 s and output is produced. To tackle this, one needs a counter that can interleave frequency measurements, so that it generates new start-points at the update rate even if the stop-point has not occurred. So, for a time-base of 10 s and an update rate of 1 s, then every 1 s a new start-trigger is produced, and then remembered until a stop trigger can be produced, at which time the start-trigger 10 s back is used to estimate the frequency. In fact, for this to work, the stop trigger time-stamp is also the start trigger time-stamp for a new measurement. You can do this with any time-base really, and the degree of interleaving only depends on the number of start-points one can keep in memory. > 2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly? There is benefits in time-measures over frequency measures when one monitors long-term properties. Also, as one tries to create a phase-curve from frequency estimates, any rounding off errors show in an slope, as there is a tiny average frequency offset from round-offs. Only really good such setups does not have significant slope. > > 3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway. The resolution of your counter tells you about where your 1/tau curve will cut tau = 1 s, and it goes from there. There is a slight scaling factor, but if we assume it is 1 for now, it is pretty simple. Your 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when the linear slope ends and "lands" in the noise, at which time the noise becomes dominant and is giving you the interesting reading. The 5370A is 20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12 at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious improvement. You are more likely to be limited by your oscillators as ref and under test at 1000 s with that one, than the instrument itself. > 4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here.  I think you have contributed by asking some really good questions. A setup I use a lot is this: 1) Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and feed into a counter Channel A/TI-start channel. For PPS signals, I make sure to trigger a but up on the rising edge not to false-trigger. For some counters this means turning of automatic trigger and set it to 1 V manually. It is important that no false triggers occurs. 2) Connect a signal under test to Channel B/TI-stop. Adjust trigger to through-zero or up on the edge as suitable. 3) TI-mode, continuous trigger 4) Collect data in TimeLab, give TimeLab the frequency of the signal on B-channel, give it the time-base of the period on the A-channel. 5) Look at data as it comes in. Look at phase view, frequency view, wrapped phase. Look at the ADEV, how the upper end flaps with data, but how the same tau becomes more and more stable as it comes in. Using even old counters this setup have helped a lot for many measures. It is simple and sturdy for many measures. Remember to save traces, to annotate it carefully so one can understand afterwards what one did. Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and then PM6853A (2 ns) to show that a particular problem did not needed the best counter in the house to be well characterized. > > 5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough? Almost. It's a good start. The first slope for lower taus is due to the instrument for sure. If you look carefully you will notice that the actual performance shifts around, because it is more complex than just being instrument limit, but it is the right ball-park for that part of the plots. For the upper end, you can be limited by your device under test drift. This can be handled by simply letting them be turned on longer. Sub-sequent measurement will have that rising slope move more and more towards higher taus and thus becomming less like a limit-issue for a certain tau. > I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics. Good spirit. DMTD takes some care, but once you learned it, it can be a magnificent tool. Cheers, Magnus > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG >  > >    On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote:  >  >  I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab) > > It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term. > Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s. > That being the case, why this consensus of what time source to use for what tau? > I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME. > I'm still trying to wrap my small head around this.  > > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >    > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow the instructions there.
MD
Magnus Danielson
Sun, Mar 1, 2020 12:20 AM

Hi,

On 2020-02-29 23:10, Taka Kamiya via time-nuts wrote:

One question for Magnus.

Ch A start - pps (standard)Ch B stop - DUT
On item 4, you said "frequency of the signal on time B".  That much is obvious.  But then you said: "give it the time-base of the period on the A-channel".  Will you explain this?
Say I give 1 Hz, period is 1s.  Say I give 10Hz, period is 0.1s.  Is this what you mean?

Yes. Exactly.

I'm using HP5370A.  This instruction is valid on this TI counter, correct??

Yes. It will work.

A request for everyone:
I am conducting an one hour measurement on HP105B.  Does anyone have 1 hour plot of this signal generator handy?  If so, will you DM me a copy?  For some reason, I cannot find one on the great Internet.

I have one of the 00105 oscillator ,as mounted and free-running in a
HP5065A, against hydrogen maser at hand. I can locate that and send you
if you wish. It's longer than 1 hour, but you get additional precision
from this.

Cheers,
Magnus

------------------- clip from Magnus's previous email------------------
A setup I use a lot is this:

  1. Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and
    feed into a counter Channel A/TI-start channel. For PPS signals, I make
    sure to trigger a but up on the rising edge not to false-trigger. For
    some counters this means turning of automatic trigger and set it to 1 V
    manually. It is important that no false triggers occurs.

  2. Connect a signal under test to Channel B/TI-stop. Adjust trigger to
    through-zero or up on the edge as suitable.

  3. TI-mode, continuous trigger

  4. Collect data in TimeLab, give TimeLab the frequency of the signal on
    B-channel, give it the time-base of the period on the A-channel.

  5. Look at data as it comes in. Look at phase view, frequency view,
    wrapped phase. Look at the ADEV, how the upper end flaps with data, but
    how the same tau becomes more and more stable as it comes in.---------------------


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

 On Friday, February 21, 2020, 9:26:47 PM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote:  

Hi Taka,

On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote:

I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point?

No problem. No worries. I hope you end up reading these and the other
replies again and acquire good knowledge. I know it's like drinking from
a fire-hoze, but you did ask some very relevant and fair questions.

1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle.

ADEV is about the frequency stability. ADEV can be calculated using
phase or frequency measures. We tend to prefer using phase measures from
Time-Interval Counters for these things.

OK, so let's say that we want to output a counter which provides output
of frequency estimates but for a time-base which is longer than 1 s,
even if we output results every 1 s?

Classically counters could not do that. You acquired a start-value,
waited the time-base, acquired a stop-value, calculated a result to
display and then arm to get a new start-value for the next result. Such
counters will have a limit that the rate of readings will be limited by
the time-base, so if it is set to 10 s, only every 10 s and output is
produced.

To tackle this, one needs a counter that can interleave frequency
measurements, so that it generates new start-points at the update rate
even if the stop-point has not occurred. So, for a time-base of 10 s and
an update rate of 1 s, then every 1 s a new start-trigger is produced,
and then remembered until a stop trigger can be produced, at which time
the start-trigger 10 s back is used to estimate the frequency. In fact,
for this to work, the stop trigger time-stamp is also the start trigger
time-stamp for a new measurement. You can do this with any time-base
really, and the degree of interleaving only depends on the number of
start-points one can keep in memory.

2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly?

There is benefits in time-measures over frequency measures when one
monitors long-term properties. Also, as one tries to create a
phase-curve from frequency estimates, any rounding off errors show in an
slope, as there is a tiny average frequency offset from round-offs. Only
really good such setups does not have significant slope.

3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway.

The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading. The 5370A is
20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12
at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious
improvement. You are more likely to be limited by your oscillators as
ref and under test at 1000 s with that one, than the instrument itself.

4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here. 

I think you have contributed by asking some really good questions.

A setup I use a lot is this:

  1. Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and
    feed into a counter Channel A/TI-start channel. For PPS signals, I make
    sure to trigger a but up on the rising edge not to false-trigger. For
    some counters this means turning of automatic trigger and set it to 1 V
    manually. It is important that no false triggers occurs.

  2. Connect a signal under test to Channel B/TI-stop. Adjust trigger to
    through-zero or up on the edge as suitable.

  3. TI-mode, continuous trigger

  4. Collect data in TimeLab, give TimeLab the frequency of the signal on
    B-channel, give it the time-base of the period on the A-channel.

  5. Look at data as it comes in. Look at phase view, frequency view,
    wrapped phase. Look at the ADEV, how the upper end flaps with data, but
    how the same tau becomes more and more stable as it comes in.

Using even old counters this setup have helped a lot for many measures.
It is simple and sturdy for many measures. Remember to save traces, to
annotate it carefully so one can understand afterwards what one did.

Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and
then PM6853A (2 ns) to show that a particular problem did not needed the
best counter in the house to be well characterized.

5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough?

Almost. It's a good start. The first slope for lower taus is due to the
instrument for sure. If you look carefully you will notice that the
actual performance shifts around, because it is more complex than just
being instrument limit, but it is the right ball-park for that part of
the plots. For the upper end, you can be limited by your device under
test drift. This can be handled by simply letting them be turned on
longer. Sub-sequent measurement will have that rising slope move more
and more towards higher taus and thus becomming less like a limit-issue
for a certain tau.

I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics.

Good spirit. DMTD takes some care, but once you learned it, it can be a
magnificent tool.

Cheers,
Magnus


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
 

    On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts time-nuts@lists.febo.com wrote: 
 
  I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab)

It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term.
Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.
That being the case, why this consensus of what time source to use for what tau?
I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME.
I'm still trying to wrap my small head around this. 


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
   


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi, On 2020-02-29 23:10, Taka Kamiya via time-nuts wrote: > One question for Magnus. > > Ch A start - pps (standard)Ch B stop - DUT > On item 4, you said "frequency of the signal on time B".  That much is obvious.  But then you said: "give it the time-base of the period on the A-channel".  Will you explain this? > Say I give 1 Hz, period is 1s.  Say I give 10Hz, period is 0.1s.  Is this what you mean? Yes. Exactly. > I'm using HP5370A.  This instruction is valid on this TI counter, correct?? Yes. It will work. > > A request for everyone: > I am conducting an one hour measurement on HP105B.  Does anyone have 1 hour plot of this signal generator handy?  If so, will you DM me a copy?  For some reason, I cannot find one on the great Internet. I have one of the 00105 oscillator ,as mounted and free-running in a HP5065A, against hydrogen maser at hand. I can locate that and send you if you wish. It's longer than 1 hour, but you get additional precision from this. Cheers, Magnus > > ------------------- clip from Magnus's previous email------------------ > A setup I use a lot is this: > 1) Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and > feed into a counter Channel A/TI-start channel. For PPS signals, I make > sure to trigger a but up on the rising edge not to false-trigger. For > some counters this means turning of automatic trigger and set it to 1 V > manually. It is important that no false triggers occurs. > > 2) Connect a signal under test to Channel B/TI-stop. Adjust trigger to > through-zero or up on the edge as suitable. > > 3) TI-mode, continuous trigger > > 4) Collect data in TimeLab, give TimeLab the frequency of the signal on > B-channel, give it the time-base of the period on the A-channel. > > 5) Look at data as it comes in. Look at phase view, frequency view, > wrapped phase. Look at the ADEV, how the upper end flaps with data, but > how the same tau becomes more and more stable as it comes in.--------------------- > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG > > > On Friday, February 21, 2020, 9:26:47 PM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote: > > Hi Taka, > > On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote: >> I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point? > No problem. No worries. I hope you end up reading these and the other > replies again and acquire good knowledge. I know it's like drinking from > a fire-hoze, but you did ask some very relevant and fair questions. >> 1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle. > ADEV is about the frequency stability. ADEV can be calculated using > phase or frequency measures. We tend to prefer using phase measures from > Time-Interval Counters for these things. > > OK, so let's say that we want to output a counter which provides output > of frequency estimates but for a time-base which is longer than 1 s, > even if we output results every 1 s? > > Classically counters could not do that. You acquired a start-value, > waited the time-base, acquired a stop-value, calculated a result to > display and then arm to get a new start-value for the next result. Such > counters will have a limit that the rate of readings will be limited by > the time-base, so if it is set to 10 s, only every 10 s and output is > produced. > > To tackle this, one needs a counter that can interleave frequency > measurements, so that it generates new start-points at the update rate > even if the stop-point has not occurred. So, for a time-base of 10 s and > an update rate of 1 s, then every 1 s a new start-trigger is produced, > and then remembered until a stop trigger can be produced, at which time > the start-trigger 10 s back is used to estimate the frequency. In fact, > for this to work, the stop trigger time-stamp is also the start trigger > time-stamp for a new measurement. You can do this with any time-base > really, and the degree of interleaving only depends on the number of > start-points one can keep in memory. > >> 2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly? > There is benefits in time-measures over frequency measures when one > monitors long-term properties. Also, as one tries to create a > phase-curve from frequency estimates, any rounding off errors show in an > slope, as there is a tiny average frequency offset from round-offs. Only > really good such setups does not have significant slope. > >> 3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway. > The resolution of your counter tells you about where your 1/tau curve > will cut tau = 1 s, and it goes from there. There is a slight scaling > factor, but if we assume it is 1 for now, it is pretty simple. Your > 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 > at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when > the linear slope ends and "lands" in the noise, at which time the noise > becomes dominant and is giving you the interesting reading. The 5370A is > 20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12 > at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious > improvement. You are more likely to be limited by your oscillators as > ref and under test at 1000 s with that one, than the instrument itself. > >> 4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here.  > I think you have contributed by asking some really good questions. > > A setup I use a lot is this: > > 1) Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and > feed into a counter Channel A/TI-start channel. For PPS signals, I make > sure to trigger a but up on the rising edge not to false-trigger. For > some counters this means turning of automatic trigger and set it to 1 V > manually. It is important that no false triggers occurs. > > 2) Connect a signal under test to Channel B/TI-stop. Adjust trigger to > through-zero or up on the edge as suitable. > > 3) TI-mode, continuous trigger > > 4) Collect data in TimeLab, give TimeLab the frequency of the signal on > B-channel, give it the time-base of the period on the A-channel. > > 5) Look at data as it comes in. Look at phase view, frequency view, > wrapped phase. Look at the ADEV, how the upper end flaps with data, but > how the same tau becomes more and more stable as it comes in. > > Using even old counters this setup have helped a lot for many measures. > It is simple and sturdy for many measures. Remember to save traces, to > annotate it carefully so one can understand afterwards what one did. > > Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and > then PM6853A (2 ns) to show that a particular problem did not needed the > best counter in the house to be well characterized. > >> 5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough? > Almost. It's a good start. The first slope for lower taus is due to the > instrument for sure. If you look carefully you will notice that the > actual performance shifts around, because it is more complex than just > being instrument limit, but it is the right ball-park for that part of > the plots. For the upper end, you can be limited by your device under > test drift. This can be handled by simply letting them be turned on > longer. Sub-sequent measurement will have that rising slope move more > and more towards higher taus and thus becomming less like a limit-issue > for a certain tau. >> I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics. > Good spirit. DMTD takes some care, but once you learned it, it can be a > magnificent tool. > > Cheers, > Magnus > >> --------------------------------------- >> (Mr.) Taka Kamiya >> KB4EMF / ex JF2DKG >>   >> >>     On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote:  >>   >>   I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab) >> >> It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term. >> Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s. >> That being the case, why this consensus of what time source to use for what tau? >> I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME. >> I'm still trying to wrap my small head around this.  >> >> --------------------------------------- >> (Mr.) Taka Kamiya >> KB4EMF / ex JF2DKG >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >>     >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
TK
Taka Kamiya
Sun, Mar 1, 2020 12:25 AM

YES, please.  tkamiya9@yahoo.com.
So far, I've taken HP105B and did adev frequency reading based, T.I. based adev, and while at it, I am doing PRS-10 T.I. based.
I have a question.  My 1 second reference for channel A is coming from GPS based 1 second.  I understand it's only 10E-8 precision on second to second basis?  Is this sufficient for OCXO and Rb based oscillators?


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG

On Saturday, February 29, 2020, 7:21:28 PM EST, Magnus Danielson <magnus@rubidium.se> wrote:  

Hi,

On 2020-02-29 23:10, Taka Kamiya via time-nuts wrote:

One question for Magnus.

Ch A start - pps (standard)Ch B stop - DUT
On item 4, you said "frequency of the signal on time B".  That much is obvious.  But then you said: "give it the time-base of the period on the A-channel".  Will you explain this?
Say I give 1 Hz, period is 1s.  Say I give 10Hz, period is 0.1s.  Is this what you mean?

Yes. Exactly.

I'm using HP5370A.  This instruction is valid on this TI counter, correct??

Yes. It will work.

A request for everyone:
I am conducting an one hour measurement on HP105B.  Does anyone have 1 hour plot of this signal generator handy?  If so, will you DM me a copy?  For some reason, I cannot find one on the great Internet.

I have one of the 00105 oscillator ,as mounted and free-running in a
HP5065A, against hydrogen maser at hand. I can locate that and send you
if you wish. It's longer than 1 hour, but you get additional precision
from this.

Cheers,
Magnus

------------------- clip from Magnus's previous email------------------
A setup I use a lot is this:

  1. Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and
    feed into a counter Channel A/TI-start channel. For PPS signals, I make
    sure to trigger a but up on the rising edge not to false-trigger. For
    some counters this means turning of automatic trigger and set it to 1 V
    manually. It is important that no false triggers occurs.

  2. Connect a signal under test to Channel B/TI-stop. Adjust trigger to
    through-zero or up on the edge as suitable.

  3. TI-mode, continuous trigger

  4. Collect data in TimeLab, give TimeLab the frequency of the signal on
    B-channel, give it the time-base of the period on the A-channel.

  5. Look at data as it comes in. Look at phase view, frequency view,
    wrapped phase. Look at the ADEV, how the upper end flaps with data, but
    how the same tau becomes more and more stable as it comes in.---------------------


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
 

    On Friday, February 21, 2020, 9:26:47 PM EST, Magnus Danielson via time-nuts time-nuts@lists.febo.com wrote: 
 
  Hi Taka,

On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote:

I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point?

No problem. No worries. I hope you end up reading these and the other
replies again and acquire good knowledge. I know it's like drinking from
a fire-hoze, but you did ask some very relevant and fair questions.

1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle.

ADEV is about the frequency stability. ADEV can be calculated using
phase or frequency measures. We tend to prefer using phase measures from
Time-Interval Counters for these things.

OK, so let's say that we want to output a counter which provides output
of frequency estimates but for a time-base which is longer than 1 s,
even if we output results every 1 s?

Classically counters could not do that. You acquired a start-value,
waited the time-base, acquired a stop-value, calculated a result to
display and then arm to get a new start-value for the next result. Such
counters will have a limit that the rate of readings will be limited by
the time-base, so if it is set to 10 s, only every 10 s and output is
produced.

To tackle this, one needs a counter that can interleave frequency
measurements, so that it generates new start-points at the update rate
even if the stop-point has not occurred. So, for a time-base of 10 s and
an update rate of 1 s, then every 1 s a new start-trigger is produced,
and then remembered until a stop trigger can be produced, at which time
the start-trigger 10 s back is used to estimate the frequency. In fact,
for this to work, the stop trigger time-stamp is also the start trigger
time-stamp for a new measurement. You can do this with any time-base
really, and the degree of interleaving only depends on the number of
start-points one can keep in memory.

2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly?

There is benefits in time-measures over frequency measures when one
monitors long-term properties. Also, as one tries to create a
phase-curve from frequency estimates, any rounding off errors show in an
slope, as there is a tiny average frequency offset from round-offs. Only
really good such setups does not have significant slope.

3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway.

The resolution of your counter tells you about where your 1/tau curve
will cut tau = 1 s, and it goes from there. There is a slight scaling
factor, but if we assume it is 1 for now, it is pretty simple. Your
5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10
at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when
the linear slope ends and "lands" in the noise, at which time the noise
becomes dominant and is giving you the interesting reading. The 5370A is
20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12
at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious
improvement. You are more likely to be limited by your oscillators as
ref and under test at 1000 s with that one, than the instrument itself.

4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here. 

I think you have contributed by asking some really good questions.

A setup I use a lot is this:

  1. Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and
    feed into a counter Channel A/TI-start channel. For PPS signals, I make
    sure to trigger a but up on the rising edge not to false-trigger. For
    some counters this means turning of automatic trigger and set it to 1 V
    manually. It is important that no false triggers occurs.

  2. Connect a signal under test to Channel B/TI-stop. Adjust trigger to
    through-zero or up on the edge as suitable.

  3. TI-mode, continuous trigger

  4. Collect data in TimeLab, give TimeLab the frequency of the signal on
    B-channel, give it the time-base of the period on the A-channel.

  5. Look at data as it comes in. Look at phase view, frequency view,
    wrapped phase. Look at the ADEV, how the upper end flaps with data, but
    how the same tau becomes more and more stable as it comes in.

Using even old counters this setup have helped a lot for many measures.
It is simple and sturdy for many measures. Remember to save traces, to
annotate it carefully so one can understand afterwards what one did.

Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and
then PM6853A (2 ns) to show that a particular problem did not needed the
best counter in the house to be well characterized.

5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough?

Almost. It's a good start. The first slope for lower taus is due to the
instrument for sure. If you look carefully you will notice that the
actual performance shifts around, because it is more complex than just
being instrument limit, but it is the right ball-park for that part of
the plots. For the upper end, you can be limited by your device under
test drift. This can be handled by simply letting them be turned on
longer. Sub-sequent measurement will have that rising slope move more
and more towards higher taus and thus becomming less like a limit-issue
for a certain tau.

I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics.

Good spirit. DMTD takes some care, but once you learned it, it can be a
magnificent tool.

Cheers,
Magnus


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG
 

    On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts time-nuts@lists.febo.com wrote: 
 
  I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab)

It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term.
Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.
That being the case, why this consensus of what time source to use for what tau?
I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME.
I'm still trying to wrap my small head around this. 


(Mr.) Taka Kamiya
KB4EMF / ex JF2DKG


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
   


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
 


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

YES, please.  tkamiya9@yahoo.com. So far, I've taken HP105B and did adev frequency reading based, T.I. based adev, and while at it, I am doing PRS-10 T.I. based. I have a question.  My 1 second reference for channel A is coming from GPS based 1 second.  I understand it's only 10E-8 precision on second to second basis?  Is this sufficient for OCXO and Rb based oscillators? --------------------------------------- (Mr.) Taka Kamiya KB4EMF / ex JF2DKG On Saturday, February 29, 2020, 7:21:28 PM EST, Magnus Danielson <magnus@rubidium.se> wrote: Hi, On 2020-02-29 23:10, Taka Kamiya via time-nuts wrote: > One question for Magnus. > > Ch A start - pps (standard)Ch B stop - DUT > On item 4, you said "frequency of the signal on time B".  That much is obvious.  But then you said: "give it the time-base of the period on the A-channel".  Will you explain this? > Say I give 1 Hz, period is 1s.  Say I give 10Hz, period is 0.1s.  Is this what you mean? Yes. Exactly. > I'm using HP5370A.  This instruction is valid on this TI counter, correct?? Yes. It will work. > > A request for everyone: > I am conducting an one hour measurement on HP105B.  Does anyone have 1 hour plot of this signal generator handy?  If so, will you DM me a copy?  For some reason, I cannot find one on the great Internet. I have one of the 00105 oscillator ,as mounted and free-running in a HP5065A, against hydrogen maser at hand. I can locate that and send you if you wish. It's longer than 1 hour, but you get additional precision from this. Cheers, Magnus > > ------------------- clip from Magnus's previous email------------------ > A setup I use a lot is this: > 1) Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and > feed into a counter Channel A/TI-start channel. For PPS signals, I make > sure to trigger a but up on the rising edge not to false-trigger. For > some counters this means turning of automatic trigger and set it to 1 V > manually. It is important that no false triggers occurs. > > 2) Connect a signal under test to Channel B/TI-stop. Adjust trigger to > through-zero or up on the edge as suitable. > > 3) TI-mode, continuous trigger > > 4) Collect data in TimeLab, give TimeLab the frequency of the signal on > B-channel, give it the time-base of the period on the A-channel. > > 5) Look at data as it comes in. Look at phase view, frequency view, > wrapped phase. Look at the ADEV, how the upper end flaps with data, but > how the same tau becomes more and more stable as it comes in.--------------------- > --------------------------------------- > (Mr.) Taka Kamiya > KB4EMF / ex JF2DKG >  > >    On Friday, February 21, 2020, 9:26:47 PM EST, Magnus Danielson via time-nuts <time-nuts@lists.febo.com> wrote:  >  >  Hi Taka, > > On 2020-02-21 23:26, Taka Kamiya via time-nuts wrote: >> I'm sorry, I messed up.  I jumped on more advance topic than I intended.  I'm sure there were answers in the replies but they must have gone way over my head because some of original questions still remain.  I bulletized (is that a word?) the original question with my NEW understanding.  Would someone please respond for me, point-to-point? > No problem. No worries. I hope you end up reading these and the other > replies again and acquire good knowledge. I know it's like drinking from > a fire-hoze, but you did ask some very relevant and fair questions. >> 1)  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s.  I understand now, Adev is about phase, not the frequency.  But assuming DUT is sine wave, if there is enough phase change, frequency do change.  I think of phase change as frequency change that is less than full cycle.  So how does counters that outputs every 1 second end up in tau of 1000s?  It will entirely miss phase change that spans more than 1 cycle. > ADEV is about the frequency stability. ADEV can be calculated using > phase or frequency measures. We tend to prefer using phase measures from > Time-Interval Counters for these things. > > OK, so let's say that we want to output a counter which provides output > of frequency estimates but for a time-base which is longer than 1 s, > even if we output results every 1 s? > > Classically counters could not do that. You acquired a start-value, > waited the time-base, acquired a stop-value, calculated a result to > display and then arm to get a new start-value for the next result. Such > counters will have a limit that the rate of readings will be limited by > the time-base, so if it is set to 10 s, only every 10 s and output is > produced. > > To tackle this, one needs a counter that can interleave frequency > measurements, so that it generates new start-points at the update rate > even if the stop-point has not occurred. So, for a time-base of 10 s and > an update rate of 1 s, then every 1 s a new start-trigger is produced, > and then remembered until a stop trigger can be produced, at which time > the start-trigger 10 s back is used to estimate the frequency. In fact, > for this to work, the stop trigger time-stamp is also the start trigger > time-stamp for a new measurement. You can do this with any time-base > really, and the degree of interleaving only depends on the number of > start-points one can keep in memory. > >> 2)  I recall reading on TICC manual, in time interval mode, anything that's reasonably good is good enough, because it has time stamp and the count reading.  Clock is used to chunk the data.  Is this still true?  Through this discussion, I ended up with conclusion that there is no inherent advantage over TI measurement when compared to frequency measurement.  Am I understanding this correctly? > There is benefits in time-measures over frequency measures when one > monitors long-term properties. Also, as one tries to create a > phase-curve from frequency estimates, any rounding off errors show in an > slope, as there is a tiny average frequency offset from round-offs. Only > really good such setups does not have significant slope. > >> 3)  I understand even the BEST counter is only good for Adev nE-12 measurement. Then, with my collection of counters, HP53132A (which averages tons of short period measurement), 5335A (not enough resolution), HP5370A (interval reading is no better than frequency), TICC by TAPR,   Do I even have a chance of doing any meaningful work?  (say work with GPSDO and Rb which some of it does reach E-13)  Yes, I know now, it is NOT possible to do 1 sec Adev but say over 100 seconds?  Right now, I don't have any standard that has adev that good at 1 sec anyway. > The resolution of your counter tells you about where your 1/tau curve > will cut tau = 1 s, and it goes from there. There is a slight scaling > factor, but if we assume it is 1 for now, it is pretty simple. Your > 5335A has 1 ns single-shot resolution, this gives 1E-9 at 1 s, but 1E-10 > at 10 s, 1E-11 at 100 s and 1E-12 at 1000 s. You see very clearly when > the linear slope ends and "lands" in the noise, at which time the noise > becomes dominant and is giving you the interesting reading. The 5370A is > 20 ps single-shot resolution, giving you a whopping 2E-11 at 1 s, 2E-12 > at 10 s, 2E-13 at 100 s and 2E-14 at 1000 s. It's some serious > improvement. You are more likely to be limited by your oscillators as > ref and under test at 1000 s with that one, than the instrument itself. > >> 4)  Would one person who has infinite patience and experience guide me through getting one reading done correctly with what I already have?  That may include email and phone call.  (I speak English and Japanese)  I don't want to lower S/N of this mailing list by doing this here.  > I think you have contributed by asking some really good questions. > > A setup I use a lot is this: > > 1) Connect a reference oscillator to produce a 1 Hz or 10 Hz signal and > feed into a counter Channel A/TI-start channel. For PPS signals, I make > sure to trigger a but up on the rising edge not to false-trigger. For > some counters this means turning of automatic trigger and set it to 1 V > manually. It is important that no false triggers occurs. > > 2) Connect a signal under test to Channel B/TI-stop. Adjust trigger to > through-zero or up on the edge as suitable. > > 3) TI-mode, continuous trigger > > 4) Collect data in TimeLab, give TimeLab the frequency of the signal on > B-channel, give it the time-base of the period on the A-channel. > > 5) Look at data as it comes in. Look at phase view, frequency view, > wrapped phase. Look at the ADEV, how the upper end flaps with data, but > how the same tau becomes more and more stable as it comes in. > > Using even old counters this setup have helped a lot for many measures. > It is simple and sturdy for many measures. Remember to save traces, to > annotate it carefully so one can understand afterwards what one did. > > Using this setup, I swapped a HP53132A (150 ps) for a HP5335A (1 ns) and > then PM6853A (2 ns) to show that a particular problem did not needed the > best counter in the house to be well characterized. > >> 5)  One time, it was mentioned many of Adev graphs posted are basically a graph of instruments noise graph.  How do I tell when a given reading/graph is exceeding the limit of a setup?  I did do base line where same signal goes to counter's reference input and signal input.  I always have that on my chart so traces does not go below.  Is that enough? > Almost. It's a good start. The first slope for lower taus is due to the > instrument for sure. If you look carefully you will notice that the > actual performance shifts around, because it is more complex than just > being instrument limit, but it is the right ball-park for that part of > the plots. For the upper end, you can be limited by your device under > test drift. This can be handled by simply letting them be turned on > longer. Sub-sequent measurement will have that rising slope move more > and more towards higher taus and thus becomming less like a limit-issue > for a certain tau. >> I appreciate everyone's input.  I am learning a lot but just not digesting well enough.  I'd like to do DMTD after I understand the basics. > Good spirit. DMTD takes some care, but once you learned it, it can be a > magnificent tool. > > Cheers, > Magnus > >> --------------------------------------- >> (Mr.) Taka Kamiya >> KB4EMF / ex JF2DKG >>   >> >>     On Thursday, February 20, 2020, 1:41:06 PM EST, Taka Kamiya via time-nuts <time-nuts@lists.febo.com> wrote:  >>   >>   I have a question concerning frequency standard and their Allen deviation.  (to measure Allen Dev in frequency mode using TimeLab) >> >> It is commonly said that for shorter tau measurement, I'd need OCXO because it's short tau jitter is superior to just about anything else.  Also, it is said that for longer tau measurement, I'd need something like Rb or Cs which has superior stability over longer term. >> Here's the question part.  A frequency counter that measures DUT basically puts out a reading every second during the measurement.  When TimeLab is well into 1000s or so, it is still reading every second; it does not change the gate time to say, 1000s. >> That being the case, why this consensus of what time source to use for what tau? >> I recall reading on TICC, in time interval mode, anything that's reasonably good is good enough.  I'm aware TI mode and Freq mode is entirely different, but it is the same in fact that measurement is made for very short time span AT A TIME. >> I'm still trying to wrap my small head around this.  >> >> --------------------------------------- >> (Mr.) Taka Kamiya >> KB4EMF / ex JF2DKG >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >>     >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >  > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow the instructions there.