volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Re: [volt-nuts] Improving Solartron 7071 and 7081 (m k)

BE
Bill Ezell
Fri, Feb 1, 2013 9:49 PM

As far as I can determine this appears to be caused by a big spike as IC403 (74LS02 on the earthy processor board) switches.  You can see the spike at TP 3 on the earthy processor board.  This spike goes  below ground by about 1.4V every 0.66uS.

This is interesting. I don't see this at all. I just probed the tp and
it's a very clean TTL signal, no undershoot at all. I'm using a 300 Mhz
scope (Tek 2465A). Not only is the tp clean, the +5 is also clean. I'm
using the ground point very near the tp.

Silly question - have you checked your probe compensation?

There is a very small rf burst every 0.66 uS I can see on the floating
board, but it's not on the digital power line, I checked that on one of
the ICs on the floating board. I'm seeing about 20 mv bursts at the AC
buffer test point you mentioned before, the burst is quite short, and
the burst frequency is much higher than any of the clocks in the meter,
so something is ringing, but it's not getting coupled out very far. I'll
probe around some more.

--
Bill Ezell

They said 'Windows or better'
so I used Linux.

> As far as I can determine this appears to be caused by a big spike as IC403 (74LS02 on the earthy processor board) switches. You can see the spike at TP 3 on the earthy processor board. This spike goes below ground by about 1.4V every 0.66uS. This is interesting. I don't see this at all. I just probed the tp and it's a very clean TTL signal, no undershoot at all. I'm using a 300 Mhz scope (Tek 2465A). Not only is the tp clean, the +5 is also clean. I'm using the ground point very near the tp. Silly question - have you checked your probe compensation? There is a very small rf burst every 0.66 uS I can see on the floating board, but it's not on the digital power line, I checked that on one of the ICs on the floating board. I'm seeing about 20 mv bursts at the AC buffer test point you mentioned before, the burst is quite short, and the burst frequency is much higher than any of the clocks in the meter, so something is ringing, but it's not getting coupled out very far. I'll probe around some more. -- Bill Ezell ---- They said 'Windows or better' so I used Linux.
DC
David C. Partridge
Sat, Feb 2, 2013 11:18 AM

I suspect probe ground lead inductance in retrospect.

As for the RF burst it matches the clock and I can pick it up with a probe with the earth lead connected to the probe tip. Pass it near the ribbon cable connecting the earthy processor board to the lower earthy logic board, or near the 74LS TTL chips on the erathy processor board and you will see it fine.

Regards,
David Partridge
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Bill Ezell
Sent: 01 February 2013 21:49
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Improving Solartron 7071 and 7081 (m k)

As far as I can determine this appears to be caused by a big spike as IC403 (74LS02 on the earthy processor board) switches.  You can see the spike at TP 3 on the earthy processor board.  This spike goes  below ground by about 1.4V every 0.66uS.

This is interesting. I don't see this at all. I just probed the tp and it's a very clean TTL signal, no undershoot at all. I'm using a 300 Mhz scope (Tek 2465A). Not only is the tp clean, the +5 is also clean. I'm using the ground point very near the tp.

Silly question - have you checked your probe compensation?

There is a very small rf burst every 0.66 uS I can see on the floating board, but it's not on the digital power line, I checked that on one of the ICs on the floating board. I'm seeing about 20 mv bursts at the AC buffer test point you mentioned before, the burst is quite short, and the burst frequency is much higher than any of the clocks in the meter, so something is ringing, but it's not getting coupled out very far. I'll probe around some more.

--
Bill Ezell

They said 'Windows or better'
so I used Linux.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I suspect probe ground lead inductance in retrospect. As for the RF burst it matches the clock and I can pick it up with a probe with the earth lead connected to the probe tip. Pass it near the ribbon cable connecting the earthy processor board to the lower earthy logic board, or near the 74LS TTL chips on the erathy processor board and you will see it fine. Regards, David Partridge -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Bill Ezell Sent: 01 February 2013 21:49 To: volt-nuts@febo.com Subject: Re: [volt-nuts] Improving Solartron 7071 and 7081 (m k) > As far as I can determine this appears to be caused by a big spike as IC403 (74LS02 on the earthy processor board) switches. You can see the spike at TP 3 on the earthy processor board. This spike goes below ground by about 1.4V every 0.66uS. This is interesting. I don't see this at all. I just probed the tp and it's a very clean TTL signal, no undershoot at all. I'm using a 300 Mhz scope (Tek 2465A). Not only is the tp clean, the +5 is also clean. I'm using the ground point very near the tp. Silly question - have you checked your probe compensation? There is a very small rf burst every 0.66 uS I can see on the floating board, but it's not on the digital power line, I checked that on one of the ICs on the floating board. I'm seeing about 20 mv bursts at the AC buffer test point you mentioned before, the burst is quite short, and the burst frequency is much higher than any of the clocks in the meter, so something is ringing, but it's not getting coupled out very far. I'll probe around some more. -- Bill Ezell ---- They said 'Windows or better' so I used Linux. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.