Hi Randy,
to modify the 845 with a modern chopper OpAmp is a great idea, and will
for sure improve its performance, rather than degrade it.
Fluke never specified the leakage currents of its chopAmps, neither in
the 845, nor in the 332 / 335 or 334 MOSFET Chopamps..
The real replacement input resistance was never specified, only the
isolation resistance, but not described, against which potential, or in
which arrangement. Compare that to FET and Chopper OpAmp specifications!
And the optical chopper of the 845 goes into bipolar transistors, not
even FETs were used!
Therefore, the resulting input currents, caused by imperfections of the
optical chopper, may also be in the high pAs, or nA, nobody ever has
measured that, but will also give arithmetical 10E13 Ohm at voltage
differences below µV level...
The great thing about the 845 is its 1kV isolation of input to monitor
output, but this will not be changed, if you replace the bipolar /
optical chopper amplifier stuff..
I think, nobody dared to modify the 845, or questioned it, because this
instrument is a "myth", indicated by the price, you still encounter on
ebay...
For the Hp419, everything i described is the same here, the difference
of this circuitry is the lack of the isolation transformers only.
Therefore, go ahead, and maybe - please specify the bias currents before
/ after modification.
Frank
Hi Frank
I have read your comments on the make-up and performance of the 845, so how
do you compare the Keithley 155 which I think was a later design ?
Roy
-----Original Message-----
From: Frank Stellmach
Sent: Tuesday, August 05, 2014 10:33 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] HP-419 and Fluke 845 Modifications
Hi Randy,
to modify the 845 with a modern chopper OpAmp is a great idea, and will
for sure improve its performance, rather than degrade it.
Fluke never specified the leakage currents of its chopAmps, neither in
the 845, nor in the 332 / 335 or 334 MOSFET Chopamps..
The real replacement input resistance was never specified, only the
isolation resistance, but not described, against which potential, or in
which arrangement. Compare that to FET and Chopper OpAmp specifications!
And the optical chopper of the 845 goes into bipolar transistors, not
even FETs were used!
Therefore, the resulting input currents, caused by imperfections of the
optical chopper, may also be in the high pAs, or nA, nobody ever has
measured that, but will also give arithmetical 10E13 Ohm at voltage
differences below µV level...
The great thing about the 845 is its 1kV isolation of input to monitor
output, but this will not be changed, if you replace the bipolar /
optical chopper amplifier stuff..
I think, nobody dared to modify the 845, or questioned it, because this
instrument is a "myth", indicated by the price, you still encounter on
ebay...
For the Hp419, everything i described is the same here, the difference
of this circuitry is the lack of the isolation transformers only.
Therefore, go ahead, and maybe - please specify the bias currents before
/ after modification.
Frank
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
I have no personal experience with the 845, however am using the keithley 155 as nullmeter fo precision calibration. I had used the 182-M (a low noise version of the 182) and other digital gear before such as the 2182, and the 155 performs very well, is easy to use with not much noise and very stable. I have seen no reason to buy a 845 or modify the 155, it is very nice as it is.
Gesendet: Mittwoch, 06. August 2014 um 04:30 Uhr
Von: "R.Phillips" phill.r1@btinternet.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] HP-419 and Fluke 845 Modifications
Hi Frank
I have read your comments on the make-up and performance of the 845, so how
do you compare the Keithley 155 which I think was a later design ?
Roy
-----Original Message-----
From: Frank Stellmach
Sent: Tuesday, August 05, 2014 10:33 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] HP-419 and Fluke 845 Modifications
Hi Randy,
to modify the 845 with a modern chopper OpAmp is a great idea, and will
for sure improve its performance, rather than degrade it.
Fluke never specified the leakage currents of its chopAmps, neither in
the 845, nor in the 332 / 335 or 334 MOSFET Chopamps..
The real replacement input resistance was never specified, only the
isolation resistance, but not described, against which potential, or in
which arrangement. Compare that to FET and Chopper OpAmp specifications!
And the optical chopper of the 845 goes into bipolar transistors, not
even FETs were used!
Therefore, the resulting input currents, caused by imperfections of the
optical chopper, may also be in the high pAs, or nA, nobody ever has
measured that, but will also give arithmetical 10E13 Ohm at voltage
differences below µV level...
The great thing about the 845 is its 1kV isolation of input to monitor
output, but this will not be changed, if you replace the bipolar /
optical chopper amplifier stuff..
I think, nobody dared to modify the 845, or questioned it, because this
instrument is a "myth", indicated by the price, you still encounter on
ebay...
For the Hp419, everything i described is the same here, the difference
of this circuitry is the lack of the isolation transformers only.
Therefore, go ahead, and maybe - please specify the bias currents before
/ after modification.
Frank
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.