Calibration and Certification - Trust and detail

RA
Robert Atkinson
Sun, Aug 11, 2013 2:39 PM

On a related issue to the 3457A calibration, I was asked to review an item at work last week (sorry can't say what or why). Looking at compliance certification by two fully qualified, internationally recognised labs, all looked well until I looked at the detail. Things like "..all RELEVANT requirements of.." (my capitals) with no list of what was relevant or any test report., details of only one test when the standard needed a number under different conditions, no record of the part or serial numbers of the unit tested. Others had accepted the "certification" at face value.  Unfortunatly if the test house "customer" says "just do this bit" and the test house is happy to put weasely words on the certifcate, then user beware. 

Robert G8RPI.

On a related issue to the 3457A calibration, I was asked to review an item at work last week (sorry can't say what or why). Looking at compliance certification by two fully qualified, internationally recognised labs, all looked well until I looked at the detail. Things like "..all RELEVANT requirements of.." (my capitals) with no list of what was relevant or any test report., details of only one test when the standard needed a number under different conditions, no record of the part or serial numbers of the unit tested. Others had accepted the "certification" at face value.  Unfortunatly if the test house "customer" says "just do this bit" and the test house is happy to put weasely words on the certifcate, then user beware.  Robert G8RPI.
DD
Dr. David Kirkby
Sun, Aug 11, 2013 9:55 PM

On 11 August 2013 15:39, Robert Atkinson robert8rpi@yahoo.co.uk wrote:

On a related issue to the 3457A calibration, I was asked to review an item at work last week (sorry can't say what or why). Looking at compliance certification by two fully qualified, internationally recognised labs, all looked well until I looked at the detail. Things like "..all RELEVANT requirements of.." (my capitals) with no list of what was relevant or any test report., details of only one test when the standard needed a number under different conditions, no record of the part or serial numbers of the unit tested. Others had accepted the "certification" at face value.

I think it depens why you want the cal certificate.

  1. If it to keep the BSI person happy to keep your ISO 9001
    acreditation, then I guess as long as it has a cal certificate that is
    ok, so use the cheapest dodgy cal lab.

  2. If the purpose of the cal certificate is to get the highest price
    when selling something, then a cal certificate by some dodgy cal lab
    is probably all you need. 99% of buyers are not going to question who
    did the cal, and their ability to do it.

  3. If you want to know the instrument works correctly, and have it
    adjusted for best performance, then it is a very different matter. One
    is probably better sending it to the manufacturer in many cases.

I want to know my VNA works properly, so that is going to Agilent this
week. The cal cost on my VNA is about 5% of what I paid for the VNA.
It would be much more difficult to justify sending my 3457A to
Agilent, when the cal cost will probably be more than what I paid for
the instrument.

How useful is this

http://www.ebay.com/itm/281149723636 ?

On the fact of it, the device would give one a reasonly high
confidence something is working readlably well. I wonder if that is
good enough for a 3457A.

Unfortunatly if the test house "customer" says "just do this bit" and the test house is happy to put weasely words on the certifcate, then user beware.

As I wrote earlier, I think there is a very cosy relationship between
cal labs and test equipment dealers. It is in both their interests to
get cal certificates on items even if they are not 100%. A T+M dealer
is not likely to use a cal lab that keeps sending items back marked
"Out of specification" or similar. A cal lab does not want to lose a
customer.

Robert G8RPI.

Dave, G8WRB.

On 11 August 2013 15:39, Robert Atkinson <robert8rpi@yahoo.co.uk> wrote: > On a related issue to the 3457A calibration, I was asked to review an item at work last week (sorry can't say what or why). Looking at compliance certification by two fully qualified, internationally recognised labs, all looked well until I looked at the detail. Things like "..all RELEVANT requirements of.." (my capitals) with no list of what was relevant or any test report., details of only one test when the standard needed a number under different conditions, no record of the part or serial numbers of the unit tested. Others had accepted the "certification" at face value. I think it depens why you want the cal certificate. 1) If it to keep the BSI person happy to keep your ISO 9001 acreditation, then I guess as long as it has a cal certificate that is ok, so use the cheapest dodgy cal lab. 2) If the purpose of the cal certificate is to get the highest price when selling something, then a cal certificate by some dodgy cal lab is probably all you need. 99% of buyers are not going to question who did the cal, and their ability to do it. 3) If you want to know the instrument works correctly, and have it adjusted for best performance, then it is a very different matter. One is probably better sending it to the manufacturer in many cases. I want to know my VNA works properly, so that is going to Agilent this week. The cal cost on my VNA is about 5% of what I paid for the VNA. It would be much more difficult to justify sending my 3457A to Agilent, when the cal cost will probably be more than what I paid for the instrument. How useful is this http://www.ebay.com/itm/281149723636 ? On the fact of it, the device would give one a reasonly high confidence something is working readlably well. I wonder if that is good enough for a 3457A. > Unfortunatly if the test house "customer" says "just do this bit" and the test house is happy to put weasely words on the certifcate, then user beware. As I wrote earlier, I think there is a very cosy relationship between cal labs and test equipment dealers. It is in both their interests to get cal certificates on items even if they are not 100%. A T+M dealer is not likely to use a cal lab that keeps sending items back marked "Out of specification" or similar. A cal lab does not want to lose a customer. > Robert G8RPI. Dave, G8WRB.
OE
Orin Eman
Sun, Aug 11, 2013 10:36 PM

On Sun, Aug 11, 2013 at 2:55 PM, Dr. David Kirkby drkirkby@gmail.comwrote:

How useful is this

http://www.ebay.com/itm/281149723636 ?

On the fact of it, the device would give one a reasonly high
confidence something is working readlably well. I wonder if that is
good enough for a 3457A.

It would give a reasonable check for gross errors.  They do claim to be
using a 3458A to cal them and give the 3458A cal certificate number.

Personally, I use a Geller Labs SVR-T for voltage checks at 10V

http://www.gellerlabs.com/Voltage%20References.htm

Joe Geller's calibrations are NIST traceable through his Fluke 732B.  A new
Agilent 34461A reads the SVR-T at about -3ppm, though it's a moving target
due to the temp-co of the '61A.  The goldenrubi supplied 3456A reads the
SVR-T at 8ppm high - inside 24 hour specs, but uncertain given the SVR-T's
claimed transfer accuracy of 5ppm.

I also have a Geller SVR that has been back twice for calibration.  The
first time back it was found to have changed by 1ppm.  The second time
back, I specified a different temperature so there was no as received
data.  The SVR is a reasonable choice if your lab stays at a constant
temperature.

I should go get the kelvin clips out and compare the 3456A against the '61A
on some 10k precision wirewound resistors I have.

Orin.

On Sun, Aug 11, 2013 at 2:55 PM, Dr. David Kirkby <drkirkby@gmail.com>wrote: > > > How useful is this > > http://www.ebay.com/itm/281149723636 ? > > On the fact of it, the device would give one a reasonly high > confidence something is working readlably well. I wonder if that is > good enough for a 3457A. > It would give a reasonable check for gross errors. They do claim to be using a 3458A to cal them and give the 3458A cal certificate number. Personally, I use a Geller Labs SVR-T for voltage checks at 10V http://www.gellerlabs.com/Voltage%20References.htm Joe Geller's calibrations are NIST traceable through his Fluke 732B. A new Agilent 34461A reads the SVR-T at about -3ppm, though it's a moving target due to the temp-co of the '61A. The goldenrubi supplied 3456A reads the SVR-T at 8ppm high - inside 24 hour specs, but uncertain given the SVR-T's claimed transfer accuracy of 5ppm. I also have a Geller SVR that has been back twice for calibration. The first time back it was found to have changed by 1ppm. The second time back, I specified a different temperature so there was no as received data. The SVR is a reasonable choice if your lab stays at a constant temperature. I should go get the kelvin clips out and compare the 3456A against the '61A on some 10k precision wirewound resistors I have. Orin.
JF
J. Forster
Mon, Aug 12, 2013 2:11 PM

David mentioned this:

http://www.ebay.com/itm/281149723636

What do people think of the device? Is it available direct from the
seller, and, if so, does anyone have the contact address?

Thanks,

-John

===============

David mentioned this: http://www.ebay.com/itm/281149723636 What do people think of the device? Is it available direct from the seller, and, if so, does anyone have the contact address? Thanks, -John ===============
Z
zbigniew169
Mon, Aug 12, 2013 4:24 PM

I don think it would be useful for serius purposes.

2013/8/12 J. Forster jfor@quikus.com

David mentioned this:

http://www.ebay.com/itm/281149723636

What do people think of the device? Is it available direct from the
seller, and, if so, does anyone have the contact address?

Thanks,

-John

===============


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Zbyszek

I don think it would be useful for serius purposes. 2013/8/12 J. Forster <jfor@quikus.com> > David mentioned this: > > http://www.ebay.com/itm/281149723636 > > What do people think of the device? Is it available direct from the > seller, and, if so, does anyone have the contact address? > > Thanks, > > -John > > =============== > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > -- Zbyszek
RA
Robert Atkinson
Mon, Aug 12, 2013 4:28 PM

The Geller Labs SVR-T is another good option. http://www.gellerlabs.com/Voltage%20References.htm

Robert G8RPI.


From: J. Forster jfor@quikus.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Monday, 12 August 2013, 15:11
Subject: [volt-nuts] Calibration  Device

David mentioned this:

http://www.ebay.com/itm/281149723636

What do people think of the device? Is it available direct from the
seller, and, if so, does anyone have the contact address?

Thanks,

-John

===============


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

The Geller Labs SVR-T is another good option. http://www.gellerlabs.com/Voltage%20References.htm Robert G8RPI. ________________________________ From: J. Forster <jfor@quikus.com> To: Discussion of precise voltage measurement <volt-nuts@febo.com> Sent: Monday, 12 August 2013, 15:11 Subject: [volt-nuts] Calibration Device David mentioned this: http://www.ebay.com/itm/281149723636 What do people think of the device? Is it available direct from the seller, and, if so, does anyone have the contact address? Thanks, -John =============== _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
OE
Orin Eman
Tue, Aug 13, 2013 4:56 AM

On Sun, Aug 11, 2013 at 3:36 PM, Orin Eman orin.eman@gmail.com wrote:

I should go get the kelvin clips out and compare the 3456A against the
'61A on some 10k precision wirewound resistors I have.

I did.  The resistors are MR102 series 0.01% 1/8W wirewound:

34461A: 10.000 82 +/- 0.000 90
3456A: 10.000 98 +/- 0.000 58

Using 90 day specs for the 61A and 90 day plus 0.0004% per month for the
56A.  Yes, the 3456A resistance specs are better than the shiny new 34461A.

I also compared the DC voltage ranges from 0.1 to 1000V and other than on
the 1KV range, results were within 10ppm.  Still, given the 3456A spec
sheet says add .12(input voltage/1000)^2 % on the 1KV range, I can't
complain; I got 999.984 on the 61A and 1000.062 on the 56A.  I'm pretty
sure that the superscript 2 in the spec meant squared, not that it matters
for 1000/1000.

Now has anyone calibrated a Fluke 343A?  The instructions in the manual are
entertaining.  They tell you to adjust the 1000V range to +/- 1mV, but the
post calibration check says that a freshly calibrated unit should be +/-
100uV at the 1000V setting.

Orin.

On Sun, Aug 11, 2013 at 3:36 PM, Orin Eman <orin.eman@gmail.com> wrote: > > I should go get the kelvin clips out and compare the 3456A against the > '61A on some 10k precision wirewound resistors I have. > I did. The resistors are MR102 series 0.01% 1/8W wirewound: 34461A: 10.000 82 +/- 0.000 90 3456A: 10.000 98 +/- 0.000 58 Using 90 day specs for the 61A and 90 day plus 0.0004% per month for the 56A. Yes, the 3456A resistance specs are better than the shiny new 34461A. I also compared the DC voltage ranges from 0.1 to 1000V and other than on the 1KV range, results were within 10ppm. Still, given the 3456A spec sheet says add .12(input voltage/1000)^2 % on the 1KV range, I can't complain; I got 999.984 on the 61A and 1000.062 on the 56A. I'm pretty sure that the superscript 2 in the spec meant squared, not that it matters for 1000/1000. Now has anyone calibrated a Fluke 343A? The instructions in the manual are entertaining. They tell you to adjust the 1000V range to +/- 1mV, but the post calibration check says that a freshly calibrated unit should be +/- 100uV at the 1000V setting. Orin.
RA
Robert Atkinson
Tue, Aug 13, 2013 6:17 AM

Hi ,
Newer does not mean better. It's not "nuts" level, but my Fluke 8060A 4.5 digit handhelds are more accurate and stable than my newer 89 IV. However most "engineers" would pick up the 89.
I agree with Dave on UK " engineers" For many years I've been a member of the very few groups of engineer in the UK who need a licence to do their job - aircraft, but even that has been diluted since EU regulations came into force. I'm also a Chartered Engineer similar to PE in the USA. As I understand it most US States require you to be a PE before you can do business as an "Engineer". In the UK it's a bloke with an oily rag and big hammer :-(
 
Robert CEng MRAeS G8RPI


From: Orin Eman orin.eman@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Tuesday, 13 August 2013, 5:56
Subject: Re: [volt-nuts] Calibration and Certification - Trust and detail

On Sun, Aug 11, 2013 at 3:36 PM, Orin Eman orin.eman@gmail.com wrote:

I should go get the kelvin clips out and compare the 3456A against the
'61A on some 10k precision wirewound resistors I have.

I did.  The resistors are MR102 series 0.01% 1/8W wirewound:

34461A: 10.000 82 +/- 0.000 90
3456A: 10.000 98 +/- 0.000 58

Using 90 day specs for the 61A and 90 day plus 0.0004% per month for the
56A.  Yes, the 3456A resistance specs are better than the shiny new 34461A.

I also compared the DC voltage ranges from 0.1 to 1000V and other than on
the 1KV range, results were within 10ppm.  Still, given the 3456A spec
sheet says add .12(input voltage/1000)^2 % on the 1KV range, I can't
complain; I got 999.984 on the 61A and 1000.062 on the 56A.  I'm pretty
sure that the superscript 2 in the spec meant squared, not that it matters
for 1000/1000.

Now has anyone calibrated a Fluke 343A?  The instructions in the manual are
entertaining.  They tell you to adjust the 1000V range to +/- 1mV, but the
post calibration check says that a freshly calibrated unit should be +/-
100uV at the 1000V setting.

Orin.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi , Newer does not mean better. It's not "nuts" level, but my Fluke 8060A 4.5 digit handhelds are more accurate and stable than my newer 89 IV. However most "engineers" would pick up the 89. I agree with Dave on UK " engineers" For many years I've been a member of the very few groups of engineer in the UK who need a licence to do their job - aircraft, but even that has been diluted since EU regulations came into force. I'm also a Chartered Engineer similar to PE in the USA. As I understand it most US States require you to be a PE before you can do business as an "Engineer". In the UK it's a bloke with an oily rag and big hammer :-(   Robert CEng MRAeS G8RPI ________________________________ From: Orin Eman <orin.eman@gmail.com> To: Discussion of precise voltage measurement <volt-nuts@febo.com> Sent: Tuesday, 13 August 2013, 5:56 Subject: Re: [volt-nuts] Calibration and Certification - Trust and detail On Sun, Aug 11, 2013 at 3:36 PM, Orin Eman <orin.eman@gmail.com> wrote: > > I should go get the kelvin clips out and compare the 3456A against the > '61A on some 10k precision wirewound resistors I have. > I did.  The resistors are MR102 series 0.01% 1/8W wirewound: 34461A: 10.000 82 +/- 0.000 90 3456A: 10.000 98 +/- 0.000 58 Using 90 day specs for the 61A and 90 day plus 0.0004% per month for the 56A.  Yes, the 3456A resistance specs are better than the shiny new 34461A. I also compared the DC voltage ranges from 0.1 to 1000V and other than on the 1KV range, results were within 10ppm.  Still, given the 3456A spec sheet says add .12(input voltage/1000)^2 % on the 1KV range, I can't complain; I got 999.984 on the 61A and 1000.062 on the 56A.  I'm pretty sure that the superscript 2 in the spec meant squared, not that it matters for 1000/1000. Now has anyone calibrated a Fluke 343A?  The instructions in the manual are entertaining.  They tell you to adjust the 1000V range to +/- 1mV, but the post calibration check says that a freshly calibrated unit should be +/- 100uV at the 1000V setting. Orin. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
OE
Orin Eman
Tue, Aug 13, 2013 7:15 AM

On Mon, Aug 12, 2013 at 9:56 PM, Orin Eman orin.eman@gmail.com wrote:

....  Still, given the 3456A spec sheet says add .12(input voltage/1000)^2
% on the 1KV range...

Make that ".012*(input voltage/1000)^2 %"... Oops.

Orin.

On Mon, Aug 12, 2013 at 9:56 PM, Orin Eman <orin.eman@gmail.com> wrote: > .... Still, given the 3456A spec sheet says add .12(input voltage/1000)^2 > % on the 1KV range... > Make that ".012*(input voltage/1000)^2 %"... Oops. Orin.