Call Agilent about 3458A calibration then tell them that you really want
your fluke calibrated and see what they will do for you.
On Tue, Dec 31, 2013 at 3:48 PM, John Phillips john.phillips0@gmail.comwrote:
Any thing you do other than record the current voltage reading will
introduce more uncertainty and increase drift unless you have a defect.
Replacing a defective part will start the aging process over again for that
part. Do not change anything if you can help it.
Agilent is a good cal lab to use as well. They will run it for a few days
and take readings a few different times and give you a report.
Most of the time the battery is not enough to keep the box hot until they
plug it in so they count on warming up a few days.
Hope that helps.
John
On Tue, Dec 31, 2013 at 3:31 PM, Chuck Harris cfharris@erols.com wrote:
The obvious reason for that suggestion is the adjustment
pot is likely to be noisy and is likely to introduce
uncertainty.
The big question that comes to my mind is if you care about
that level of uncertainty, why not replace the pot with a
voltage divider made of some low thermal coefficient fixed
value resistors?
-Chuck Harris
jeffhook@comcast.net wrote:
Hi,
Are you guys adjusting your 732As?
I have a friend that works in a large CAL lab and he told me to never
adjust my
732As, just compare them.
They send out their 732 Bs to Fluke but are never adjusted, just
compared and sent
back with current voltage reading ??
Thanks - Jeff
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
--
John Phillips
--
John Phillips
Interesting thought. I love it when guys of that caliber compete.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of John Phillips
Sent: Tuesday, December 31, 2013 5:51 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A Questions
Call Agilent about 3458A calibration then tell them that you really want
your fluke calibrated and see what they will do for you.
On Tue, Dec 31, 2013 at 3:48 PM, John Phillips
john.phillips0@gmail.comwrote:
Any thing you do other than record the current voltage reading will
introduce more uncertainty and increase drift unless you have a defect.
Replacing a defective part will start the aging process over again for
that part. Do not change anything if you can help it.
Agilent is a good cal lab to use as well. They will run it for a few
days and take readings a few different times and give you a report.
Most of the time the battery is not enough to keep the box hot until
they plug it in so they count on warming up a few days.
Hope that helps.
John
On Tue, Dec 31, 2013 at 3:31 PM, Chuck Harris cfharris@erols.com wrote:
The obvious reason for that suggestion is the adjustment pot is
likely to be noisy and is likely to introduce uncertainty.
The big question that comes to my mind is if you care about that
level of uncertainty, why not replace the pot with a voltage divider
made of some low thermal coefficient fixed value resistors?
-Chuck Harris
jeffhook@comcast.net wrote:
Hi,
Are you guys adjusting your 732As?
I have a friend that works in a large CAL lab and he told me to
never adjust my 732As, just compare them.
They send out their 732 Bs to Fluke but are never adjusted, just
compared and sent back with current voltage reading ??
Thanks - Jeff
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/ mailman/listinfo/volt-nuts and follow
the instructions there.
--
John Phillips
--
John Phillips
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.