53230A TIC and TimeLab

A(
AC0XU (Jim)
Thu, Oct 4, 2018 2:04 AM

I have recently acquited a 53230A counter and have tried to measure ADEV of a GPSDO with it. I am suspicious of the ADEV numbers reported by TimeLab V1.31 because

  1. Timelab disagrees significantly with the ADEV numbers reported by the 53230A itself.  In this case, with 1 sec sample intervals, the 53230A reports an ADEV of about 370 micro Hz (or 3.7e-11). Time lab reports  1.5e-11 @0.01 sec sample period, 2e-11 @0.1 sec sample period, 6e-11 @1sec sample period - all at tau=1sec.

  2. When I use TimeLab to select different sample intervals, the whole ADEV curve shifts left (with smaller sample intervals) or right (with larger sample intervals).

Its as if TimeLab does not correctly account for the sample interval, but am I just not using the Timelab/counter setup correctly?

Thanks!

I have recently acquited a 53230A counter and have tried to measure ADEV of a GPSDO with it. I am suspicious of the ADEV numbers reported by TimeLab V1.31 because 1) Timelab disagrees significantly with the ADEV numbers reported by the 53230A itself. In this case, with 1 sec sample intervals, the 53230A reports an ADEV of about 370 micro Hz (or 3.7e-11). Time lab reports 1.5e-11 @0.01 sec sample period, 2e-11 @0.1 sec sample period, 6e-11 @1sec sample period - all at tau=1sec. 2) When I use TimeLab to select different sample intervals, the whole ADEV curve shifts left (with smaller sample intervals) or right (with larger sample intervals). Its as if TimeLab does not correctly account for the sample interval, but am I just not using the Timelab/counter setup correctly? Thanks!
OP
Ole Petter Rønningen
Thu, Oct 4, 2018 6:26 AM

You may want to have a look at http://www.efos3.com/53230A/HPAK53230A-1.html

Ole

  1. okt. 2018 kl. 04:04 skrev AC0XU (Jim) James.Schatzman@ac0xu.com:

I have recently acquited a 53230A counter and have tried to measure ADEV of a GPSDO with it. I am suspicious of the ADEV numbers reported by TimeLab V1.31 because

  1. Timelab disagrees significantly with the ADEV numbers reported by the 53230A itself.  In this case, with 1 sec sample intervals, the 53230A reports an ADEV of about 370 micro Hz (or 3.7e-11). Time lab reports  1.5e-11 @0.01 sec sample period, 2e-11 @0.1 sec sample period, 6e-11 @1sec sample period - all at tau=1sec.

  2. When I use TimeLab to select different sample intervals, the whole ADEV curve shifts left (with smaller sample intervals) or right (with larger sample intervals).

Its as if TimeLab does not correctly account for the sample interval, but am I just not using the Timelab/counter setup correctly?

Thanks!


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

You may want to have a look at http://www.efos3.com/53230A/HPAK53230A-1.html Ole > 4. okt. 2018 kl. 04:04 skrev AC0XU (Jim) <James.Schatzman@ac0xu.com>: > > I have recently acquited a 53230A counter and have tried to measure ADEV of a GPSDO with it. I am suspicious of the ADEV numbers reported by TimeLab V1.31 because > > 1) Timelab disagrees significantly with the ADEV numbers reported by the 53230A itself. In this case, with 1 sec sample intervals, the 53230A reports an ADEV of about 370 micro Hz (or 3.7e-11). Time lab reports 1.5e-11 @0.01 sec sample period, 2e-11 @0.1 sec sample period, 6e-11 @1sec sample period - all at tau=1sec. > > 2) When I use TimeLab to select different sample intervals, the whole ADEV curve shifts left (with smaller sample intervals) or right (with larger sample intervals). > > Its as if TimeLab does not correctly account for the sample interval, but am I just not using the Timelab/counter setup correctly? > > Thanks! > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
OP
Ole Petter Ronningen
Thu, Oct 4, 2018 6:56 AM

Somewhat longer answer; I assume you have set up a frequency measurement
(as opposed to time interval)

  1. Timelab uses SCPI-command "READ" to fetch readings from the 53230a (and
    most other counters *). The 53230a does not return gap free frequency
    measurements in this case. There will be deadtime which will be ignored, as
    you say, but the measurements will also be biased. In addition, the
    CONTinous gap free mode is broken when it comes to ADEV, as has been
    documented elsewhere. From my experience, using the 53230a in a time
    interval measurement takes care of these issues, at the expense of a
    slighly more cumbersome setup.

  2. You may know this, but from your text it could be a misunderstanding:
    timelab does not configure the instrument; whichever sample interval you
    select in timelab should match the sample interval you have set up the
    53230a to use.

  3. Regarding the ADEV number discrepancy; are the measurements made at the
    same time
    ? For "long enough"? (The numbers jump all over the place before
    they settle down, thats normal from my experience.)

Of course, theres the question of what reference you are measuring against.

Ole

  • at least this was the case the last  time I checked, but I have not
    looked at the source for 1.31 to verify.

On Thu, Oct 4, 2018 at 7:42 AM AC0XU (Jim) James.Schatzman@ac0xu.com
wrote:

I have recently acquited a 53230A counter and have tried to measure ADEV
of a GPSDO with it. I am suspicious of the ADEV numbers reported by TimeLab
V1.31 because

  1. Timelab disagrees significantly with the ADEV numbers reported by the
    53230A itself.  In this case, with 1 sec sample intervals, the 53230A
    reports an ADEV of about 370 micro Hz (or 3.7e-11). Time lab reports
    1.5e-11 @0.01 sec sample period, 2e-11 @0.1 sec sample period, 6e-11 @1sec
    sample period - all at tau=1sec.

  2. When I use TimeLab to select different sample intervals, the whole ADEV
    curve shifts left (with smaller sample intervals) or right (with larger
    sample intervals).

Its as if TimeLab does not correctly account for the sample interval, but
am I just not using the Timelab/counter setup correctly?

Thanks!


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Somewhat longer answer; I assume you have set up a frequency measurement (as opposed to time interval) 1. Timelab uses SCPI-command "READ" to fetch readings from the 53230a (and most other counters *). The 53230a does not return gap free frequency measurements in this case. There will be deadtime which will be ignored, as you say, but the measurements will also be biased. In addition, the CONTinous gap free mode is broken when it comes to ADEV, as has been documented elsewhere. From my experience, using the 53230a in a time interval measurement takes care of these issues, at the expense of a slighly more cumbersome setup. 2. You may know this, but from your text it could be a misunderstanding: timelab does not configure the instrument; whichever sample interval you select in timelab should match the sample interval you have set up the 53230a to use. 3. Regarding the ADEV number discrepancy; are the measurements made *at the same time*? For "long enough"? (The numbers jump all over the place before they settle down, thats normal from my experience.) Of course, theres the question of what reference you are measuring against. Ole * at least this was the case the last time I checked, but I have not looked at the source for 1.31 to verify. On Thu, Oct 4, 2018 at 7:42 AM AC0XU (Jim) <James.Schatzman@ac0xu.com> wrote: > I have recently acquited a 53230A counter and have tried to measure ADEV > of a GPSDO with it. I am suspicious of the ADEV numbers reported by TimeLab > V1.31 because > > 1) Timelab disagrees significantly with the ADEV numbers reported by the > 53230A itself. In this case, with 1 sec sample intervals, the 53230A > reports an ADEV of about 370 micro Hz (or 3.7e-11). Time lab reports > 1.5e-11 @0.01 sec sample period, 2e-11 @0.1 sec sample period, 6e-11 @1sec > sample period - all at tau=1sec. > > 2) When I use TimeLab to select different sample intervals, the whole ADEV > curve shifts left (with smaller sample intervals) or right (with larger > sample intervals). > > Its as if TimeLab does not correctly account for the sample interval, but > am I just not using the Timelab/counter setup correctly? > > Thanks! > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
MD
Magnus Danielson
Thu, Oct 4, 2018 8:58 AM

Hi,

On 10/4/18 8:56 AM, Ole Petter Ronningen wrote:

Somewhat longer answer; I assume you have set up a frequency measurement
(as opposed to time interval)

  1. Timelab uses SCPI-command "READ" to fetch readings from the 53230a (and
    most other counters *). The 53230a does not return gap free frequency
    measurements in this case. There will be deadtime which will be ignored, as
    you say, but the measurements will also be biased. In addition, the
    CONTinous gap free mode is broken when it comes to ADEV, as has been
    documented elsewhere. From my experience, using the 53230a in a time
    interval measurement takes care of these issues, at the expense of a
    slighly more cumbersome setup.

Regardless, the 53230A does a filtered frequency estimation, which is
great for achieving high precision frequency measures quickly, but not
good for ADEV measures, as it introduces a bias due to the lower
bandwidth of the filtering as compared to the sampling interval.
Dead-time gaps biases also comes in.

For any software (TimeLab or Stable32) to process sufficiently unbiased
values, only phase values should be used to avoid both the filtering and
dead-time gap issues.

  1. You may know this, but from your text it could be a misunderstanding:
    timelab does not configure the instrument; whichever sample interval you
    select in timelab should match the sample interval you have set up the
    53230a to use.

TimeLab actually assumes that the user really know what they are doing,
but attempts to assist in the overall task. Adding a instrument
configuration tool into TimeLab would be possible, but for most part
consider the configuration you do for a measurement the opportunity to
input to TimeLab under which conditions the measurement was done, how
the instrument was setup etc. From there it just consumes measurements,
scales them according to the configuration, unwrapp phase etc. prior to
record the measurement record that is then used for analysis.

  1. Regarding the ADEV number discrepancy; are the measurements made at the
    same time
    ? For "long enough"? (The numbers jump all over the place before
    they settle down, thats normal from my experience.)

One of the great learning experiences you can get from TimeLab is how
the real-time update of ADEV swings in the highest tau range of the
plot, but as more samples comes in, the same tau range becomes less
uncertain, the confidence bounds (chi-square based) goes down and
TimeLab illustrates this very nicely. Thus, one wants to have enough
samples before judging too much of a value. It's experience one has to
built.

Of course, theres the question of what reference you are measuring against.

Always. It is always good to look at specs and other measurements to see
if the measurement is reasonably in compliance with what one can expect
from other measurements. Even with very good instruments that does a lot
to solve the measurement problem, you always needs to validate it and
check it. One need to understand how the measurement instrument and
processing will impact the value and understand what part of this is
needed to get "proper" value and what is various forms of impairments.

Cheers,
Magnus

Ole

  • at least this was the case the last  time I checked, but I have not
    looked at the source for 1.31 to verify.

On Thu, Oct 4, 2018 at 7:42 AM AC0XU (Jim) James.Schatzman@ac0xu.com
wrote:

I have recently acquited a 53230A counter and have tried to measure ADEV
of a GPSDO with it. I am suspicious of the ADEV numbers reported by TimeLab
V1.31 because

  1. Timelab disagrees significantly with the ADEV numbers reported by the
    53230A itself.  In this case, with 1 sec sample intervals, the 53230A
    reports an ADEV of about 370 micro Hz (or 3.7e-11). Time lab reports
    1.5e-11 @0.01 sec sample period, 2e-11 @0.1 sec sample period, 6e-11 @1sec
    sample period - all at tau=1sec.

  2. When I use TimeLab to select different sample intervals, the whole ADEV
    curve shifts left (with smaller sample intervals) or right (with larger
    sample intervals).

Its as if TimeLab does not correctly account for the sample interval, but
am I just not using the Timelab/counter setup correctly?

Thanks!


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi, On 10/4/18 8:56 AM, Ole Petter Ronningen wrote: > Somewhat longer answer; I assume you have set up a frequency measurement > (as opposed to time interval) > > 1. Timelab uses SCPI-command "READ" to fetch readings from the 53230a (and > most other counters *). The 53230a does not return gap free frequency > measurements in this case. There will be deadtime which will be ignored, as > you say, but the measurements will also be biased. In addition, the > CONTinous gap free mode is broken when it comes to ADEV, as has been > documented elsewhere. From my experience, using the 53230a in a time > interval measurement takes care of these issues, at the expense of a > slighly more cumbersome setup. Regardless, the 53230A does a filtered frequency estimation, which is great for achieving high precision frequency measures quickly, but not good for ADEV measures, as it introduces a bias due to the lower bandwidth of the filtering as compared to the sampling interval. Dead-time gaps biases also comes in. For any software (TimeLab or Stable32) to process sufficiently unbiased values, only phase values should be used to avoid both the filtering and dead-time gap issues. > 2. You may know this, but from your text it could be a misunderstanding: > timelab does not configure the instrument; whichever sample interval you > select in timelab should match the sample interval you have set up the > 53230a to use. TimeLab actually assumes that the user really know what they are doing, but attempts to assist in the overall task. Adding a instrument configuration tool into TimeLab would be possible, but for most part consider the configuration you do for a measurement the opportunity to input to TimeLab under which conditions the measurement was done, how the instrument was setup etc. From there it just consumes measurements, scales them according to the configuration, unwrapp phase etc. prior to record the measurement record that is then used for analysis. > 3. Regarding the ADEV number discrepancy; are the measurements made *at the > same time*? For "long enough"? (The numbers jump all over the place before > they settle down, thats normal from my experience.) One of the great learning experiences you can get from TimeLab is how the real-time update of ADEV swings in the highest tau range of the plot, but as more samples comes in, the same tau range becomes less uncertain, the confidence bounds (chi-square based) goes down and TimeLab illustrates this very nicely. Thus, one wants to have enough samples before judging too much of a value. It's experience one has to built. > Of course, theres the question of what reference you are measuring against. Always. It is always good to look at specs and other measurements to see if the measurement is reasonably in compliance with what one can expect from other measurements. Even with very good instruments that does a lot to solve the measurement problem, you always needs to validate it and check it. One need to understand how the measurement instrument and processing will impact the value and understand what part of this is needed to get "proper" value and what is various forms of impairments. Cheers, Magnus > Ole > > * at least this was the case the last time I checked, but I have not > looked at the source for 1.31 to verify. > > On Thu, Oct 4, 2018 at 7:42 AM AC0XU (Jim) <James.Schatzman@ac0xu.com> > wrote: > >> I have recently acquited a 53230A counter and have tried to measure ADEV >> of a GPSDO with it. I am suspicious of the ADEV numbers reported by TimeLab >> V1.31 because >> >> 1) Timelab disagrees significantly with the ADEV numbers reported by the >> 53230A itself. In this case, with 1 sec sample intervals, the 53230A >> reports an ADEV of about 370 micro Hz (or 3.7e-11). Time lab reports >> 1.5e-11 @0.01 sec sample period, 2e-11 @0.1 sec sample period, 6e-11 @1sec >> sample period - all at tau=1sec. >> >> 2) When I use TimeLab to select different sample intervals, the whole ADEV >> curve shifts left (with smaller sample intervals) or right (with larger >> sample intervals). >> >> Its as if TimeLab does not correctly account for the sample interval, but >> am I just not using the Timelab/counter setup correctly? >> >> Thanks! >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
OP
Ole Petter Ronningen
Thu, Oct 4, 2018 9:21 AM

On Thu, Oct 4, 2018 at 11:00 AM Magnus Danielson magnus@rubidium.dyndns.org
wrote:

Regardless, the 53230A does a filtered frequency estimation, which is
great for achieving high precision frequency measures quickly, but not
good for ADEV measures, as it introduces a bias due to the lower
bandwidth of the filtering as compared to the sampling interval.
Dead-time gaps biases also comes in.

True - but be aware that the first sample returned from the instrument,
when fetched using "READ", will also be biased in an absolute sense - the
frequency reported will systematically be too high or too low (the
magnitude and direction of the error depending primarily on gate time, from
what I've seen).

So even when using this mode to only look at high resolution "absolute
frequency measurements", it is easy to get bitten if the data is collected
using a computer.

BR,
Ole

On Thu, Oct 4, 2018 at 11:00 AM Magnus Danielson <magnus@rubidium.dyndns.org> wrote: > Regardless, the 53230A does a filtered frequency estimation, which is > great for achieving high precision frequency measures quickly, but not > good for ADEV measures, as it introduces a bias due to the lower > bandwidth of the filtering as compared to the sampling interval. > Dead-time gaps biases also comes in. > True - but be aware that the *first sample* returned from the instrument, when fetched using "READ", will also be biased in an absolute sense - the frequency reported will systematically be too high or too low (the magnitude and direction of the error depending primarily on gate time, from what I've seen). So even when using this mode to only look at high resolution "absolute frequency measurements", it is easy to get bitten if the data is collected using a computer. BR, Ole
MD
Magnus Danielson
Thu, Oct 4, 2018 11:39 AM

Hej Ole!

On 10/4/18 11:21 AM, Ole Petter Ronningen wrote:

On Thu, Oct 4, 2018 at 11:00 AM Magnus Danielson magnus@rubidium.dyndns.org
wrote:

Regardless, the 53230A does a filtered frequency estimation, which is
great for achieving high precision frequency measures quickly, but not
good for ADEV measures, as it introduces a bias due to the lower
bandwidth of the filtering as compared to the sampling interval.
Dead-time gaps biases also comes in.

True - but be aware that the first sample returned from the instrument,
when fetched using "READ", will also be biased in an absolute sense - the
frequency reported will systematically be too high or too low (the
magnitude and direction of the error depending primarily on gate time, from
what I've seen).

Who-oh! OK. This means that one should intentionally let a number of
samples pass before trusting it.

It would be interesting to get some better qualifications of this.

So even when using this mode to only look at high resolution "absolute
frequency measurements", it is easy to get bitten if the data is collected
using a computer.

Obviously. Good to know. Many thanks!

It is interesting to learn that we still need to handle the temperament
of instruments.

Cheers,
Magnus

Hej Ole! On 10/4/18 11:21 AM, Ole Petter Ronningen wrote: > On Thu, Oct 4, 2018 at 11:00 AM Magnus Danielson <magnus@rubidium.dyndns.org> > wrote: > >> Regardless, the 53230A does a filtered frequency estimation, which is >> great for achieving high precision frequency measures quickly, but not >> good for ADEV measures, as it introduces a bias due to the lower >> bandwidth of the filtering as compared to the sampling interval. >> Dead-time gaps biases also comes in. >> > > True - but be aware that the *first sample* returned from the instrument, > when fetched using "READ", will also be biased in an absolute sense - the > frequency reported will systematically be too high or too low (the > magnitude and direction of the error depending primarily on gate time, from > what I've seen). Who-oh! OK. This means that one should intentionally let a number of samples pass before trusting it. It would be interesting to get some better qualifications of this. > So even when using this mode to only look at high resolution "absolute > frequency measurements", it is easy to get bitten if the data is collected > using a computer. Obviously. Good to know. Many thanks! It is interesting to learn that we still need to handle the temperament of instruments. Cheers, Magnus
OP
Ole Petter Ronningen
Thu, Oct 4, 2018 1:42 PM

On Thu, Oct 4, 2018 at 1:40 PM Magnus Danielson magnus@rubidium.dyndns.org
wrote:

Who-oh! OK. This means that one should intentionally let a number of
samples pass before trusting it.

Precisely. But, crucially, discard 1-2 samples since the last INIT -
which may not be obvious without some investigation. It is a bit
convoluted, but the behaviour is easily observed using TimeLab: Feed the
counter the same 10  MHz signal both on the EXTernal REFerence, and channel

  1. Set up a frequency measurement, and collect data using timelab. Observe
    the phaseplot (taking care to NOT have the phase plot in "residual mode").

After a few minutes, a slope will be observed on the phase plot that should
not be there - it is measuring its own reference, so the plot should be
pretty much dead flat over a sufficiently long interval. It is because
timelab calls READ every time it wants a sample, and the instrument returns
a biased frequency estimate. Shorter gate-times, steeper slope.

It would be interesting to get some better qualifications of this.

I have attempted to make a thorough writeup on the previously mentioned
http://www.efos3.com/53230A/HPAK53230A-1.html Also parts 2 and 3 - in
short, the behaviour has been observed on three separate instruments. The
data is available for download on my website. Oh, and Keysight has
acknowledged the issue, but not offered anything towards a solution.

Ole

On Thu, Oct 4, 2018 at 1:40 PM Magnus Danielson <magnus@rubidium.dyndns.org> wrote: > Who-oh! OK. This means that one should intentionally let a number of > samples pass before trusting it. > Precisely. But, crucially, discard 1-2 samples *since the last INIT* - which may not be obvious without some investigation. It is a bit convoluted, but the behaviour is easily observed using TimeLab: Feed the counter the same 10 MHz signal both on the EXTernal REFerence, and channel 1. Set up a frequency measurement, and collect data using timelab. Observe the phaseplot (taking care to NOT have the phase plot in "residual mode"). After a few minutes, a slope will be observed on the phase plot that should not be there - it is measuring its own reference, so the plot should be pretty much dead flat over a sufficiently long interval. It is because timelab calls READ every time it wants a sample, and the instrument returns a biased frequency estimate. Shorter gate-times, steeper slope. It would be interesting to get some better qualifications of this. I have attempted to make a thorough writeup on the previously mentioned http://www.efos3.com/53230A/HPAK53230A-1.html Also parts 2 and 3 - in short, the behaviour has been observed on three separate instruments. The data is available for download on my website. Oh, and Keysight has acknowledged the issue, but not offered anything towards a solution. Ole
BK
Bob kb8tq
Thu, Oct 4, 2018 2:02 PM

Hi

Probably also worth mentioning:

For short tau, you are probably measuring the ADEV of the counter noise floor rather than the
ADEV of the GPSDO. Yes, there are noisy GPSDO’s out there but most of what you run into
is not in that category.

Bob

On Oct 4, 2018, at 8:42 AM, Ole Petter Ronningen opronningen@gmail.com wrote:

On Thu, Oct 4, 2018 at 1:40 PM Magnus Danielson magnus@rubidium.dyndns.org
wrote:

Who-oh! OK. This means that one should intentionally let a number of
samples pass before trusting it.

Precisely. But, crucially, discard 1-2 samples since the last INIT -
which may not be obvious without some investigation. It is a bit
convoluted, but the behaviour is easily observed using TimeLab: Feed the
counter the same 10  MHz signal both on the EXTernal REFerence, and channel

  1. Set up a frequency measurement, and collect data using timelab. Observe
    the phaseplot (taking care to NOT have the phase plot in "residual mode").

After a few minutes, a slope will be observed on the phase plot that should
not be there - it is measuring its own reference, so the plot should be
pretty much dead flat over a sufficiently long interval. It is because
timelab calls READ every time it wants a sample, and the instrument returns
a biased frequency estimate. Shorter gate-times, steeper slope.

It would be interesting to get some better qualifications of this.

I have attempted to make a thorough writeup on the previously mentioned
http://www.efos3.com/53230A/HPAK53230A-1.html Also parts 2 and 3 - in
short, the behaviour has been observed on three separate instruments. The
data is available for download on my website. Oh, and Keysight has
acknowledged the issue, but not offered anything towards a solution.

Ole


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Probably also worth mentioning: For short tau, you are probably measuring the ADEV of the counter noise floor rather than the ADEV of the GPSDO. Yes, there are noisy GPSDO’s out there but most of what you run into is not in that category. Bob > On Oct 4, 2018, at 8:42 AM, Ole Petter Ronningen <opronningen@gmail.com> wrote: > > On Thu, Oct 4, 2018 at 1:40 PM Magnus Danielson <magnus@rubidium.dyndns.org> > wrote: > >> Who-oh! OK. This means that one should intentionally let a number of >> samples pass before trusting it. >> > > Precisely. But, crucially, discard 1-2 samples *since the last INIT* - > which may not be obvious without some investigation. It is a bit > convoluted, but the behaviour is easily observed using TimeLab: Feed the > counter the same 10 MHz signal both on the EXTernal REFerence, and channel > 1. Set up a frequency measurement, and collect data using timelab. Observe > the phaseplot (taking care to NOT have the phase plot in "residual mode"). > > After a few minutes, a slope will be observed on the phase plot that should > not be there - it is measuring its own reference, so the plot should be > pretty much dead flat over a sufficiently long interval. It is because > timelab calls READ every time it wants a sample, and the instrument returns > a biased frequency estimate. Shorter gate-times, steeper slope. > > It would be interesting to get some better qualifications of this. > > > I have attempted to make a thorough writeup on the previously mentioned > http://www.efos3.com/53230A/HPAK53230A-1.html Also parts 2 and 3 - in > short, the behaviour has been observed on three separate instruments. The > data is available for download on my website. Oh, and Keysight has > acknowledged the issue, but not offered anything towards a solution. > > Ole > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.