Hi,
the current temperature relationship should be nearly linear,
so if 100mA is needed at 20 C ambient and it follows that at 80 C
no heating is required, then the heating current will decrease at
about 1.7mA/K.
cheers, Neville Michie
On 23/11/2007, at 9:33 AM, Didier Juges wrote:
Bruce,
I have some 4 terminal current sense resistors at work, but they
are very
low values (0.1 ohm and lower), so what I may gain at the sensor I
may loose
in the amplifier. Right now, the 1 ohm resistor and OP-27 give a
voltage
that is way out of the noise, as the curve shows. If I can find more
temperature stable resistors for the instrument amplifier, then I
could use
one of those 4 terminal resistors. I will see what I can get to
improve this
setup. As it is, it may not be extremely accurate in absolute
terms, but it
is precise and probably sufficiently relatively accurate for what I
want to
do.
What do you think of the 2mA/degree current sensitivity? Is it in the
ballpark? I need to run the test much longer, and over greater
temperature
variations, but I am still tinkering with it... The engineer's
curse :-)
Didier
-----Original Message-----
From: time-nuts-bounces@febo.com
[mailto:time-nuts-bounces@febo.com] On Behalf Of Bruce Griffiths
Sent: Thursday, November 22, 2007 3:28 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP10811 current versus temperature
Ideally you should use a 4 terminal current sensing resistor
and a high input impedance instrumentation amplifier.
A Thomas style 4 terminal resistor in a temperature
controlled oil bath would be nice, but failing that a low
tempco 4 terminal resistor will suffice.
Bruce
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/
time-nuts
and follow the instructions there.
Didier Juges wrote:
Bruce,
I have some 4 terminal current sense resistors at work, but they are very
low values (0.1 ohm and lower), so what I may gain at the sensor I may loose
in the amplifier. Right now, the 1 ohm resistor and OP-27 give a voltage
that is way out of the noise, as the curve shows. If I can find more
temperature stable resistors for the instrument amplifier, then I could use
one of those 4 terminal resistors. I will see what I can get to improve this
setup. As it is, it may not be extremely accurate in absolute terms, but it
is precise and probably sufficiently relatively accurate for what I want to
do.
What do you think of the 2mA/degree current sensitivity? Is it in the
ballpark? I need to run the test much longer, and over greater temperature
variations, but I am still tinkering with it... The engineer's curse :-)
Didier
Didier
Its hard to find temperature stable 4 terminal resistors with tempcos
better than 50ppm/C or resistances greater than 0.1 ohm off the shelf.
The only source of 1 ohm 4 terminal resistors would appear to be used
L&N, JRL, etc 4 terminal (Thomas style) standards which can be expensive
($US2000 or so).
However have found listing for NBS pattern standard resistors at $US165.
Such standard resistors have very low tempcos but are intended for oil
bath operation.
The only current US manufacturer of such standard resistors (including
ones rated for operation in air) appears to be Ohm labs:
http://www.ohm-labs.com http://www.ohmlabs.com
Leeds and Northrup and Julie Research labs once made such resistors.
All that can be done to check that the temperature dependence of the
oven current is in the ballpark is compare your results with those in
the HP journal article on the 10811A
From figure 4 on page 21 of the March 1981 issue.
OVEN POWER
5.13W @ -50C
0.26W @ +70C
Or 40mW/C
with a 24V oven supply this corresponds to about 1.6mA/C.
So you would appear to be in the ball park.
Note above figures are for moving air somewhat less in still air.
Bruce