More Noise Floor Nonsense

BK
Bob kb8tq
Wed, Feb 10, 2021 5:18 PM

Hi

Back a few years ago, Keysight did one of their PR videos:

https://www.youtube.com/watch?v=4DfLkxVe7Lk https://www.youtube.com/watch?v=4DfLkxVe7Lk

It shows a fairly simple setup to see what your two channel
counter is doing. A lot of counters have a built in standard
deviation function so it applies to more than just the counter
he’s showing. People have been doing this sort of test for many
decades.

The one thing he’s not to clear on is what sort of signal levels
are being used. “Square wave” is about as specific as he gets.

I decided to try the same test with a 100 MHz sine wave at
2.5 dbm. Sine wave is nice since it has a well defined slope
at the zero crossing ( which does matter in this test). It also
happened to be an available output on the FS-740.

If you repeat the YouTube test with this signal, you get the same
13 ps one sigma and 9 ps resolution on a 53230A. That suggests
to me that this is a “fast enough” signal to be useful. Same test
gave the same results on a couple of 53230’s.

So … what happens if we try this on a few other counters?

A 53132A comes in right around 140 ps one sigma to give it
a 100 ps resolution. That’s nice compared to a 150 ps spec.

The “surprise” candidate is the 53220A. One would think that
with a 100 ps resolution spec, it would be 5X worse than the
53230A. Not so much. It comes in at 14 ps one sigma and 10 ps
resolution. ( Usual disclaimers apply …. used counter of
unknown (China / eBay) origin  …. one sample tested…. YMMV ….).

Does this test tell you everything about your counter? No
of course it doesn’t. There are a lot of bits and pieces in there,
this is only looking at some of them. Does it tell you what you
will see on a 10 MHz sine wave at 10 dbm? Nope, that’s a
slower edge. The one sigma is roughly 24 ps on a 53230A.

What it does provide is a quick check of how your counter
is doing. Most of what gets killed in a counter is in the input
channels (as noted in a previous post). This gives you a fast
check on how they are doing. It also checks a bit more than
that ….

To be really useful, you need to run the test with your source
on your counter(s). (It does not have to be 100 MHz and 2.5
dbm) Note the number(s) you get and file them away. Come back
a year or three later scratching your head about some readings
….. you have something to check against. Checking against spec
… not so much ….

The key thing in this test as shown in the video is the fast edge.
For a lot of years many of us have simply been using a 10 MHz
test signal. There’s nothing wrong with doing that. One might
guess that’s what the designers of the 53230 did. Turns out
that a faster edge is a better edge, even at RF frequencies.
Who knew …. ( other than anybody who watches YouTube …. :) )

Could you extend this test by playing with various levels to come
up with a good idea of what to expect at this or that level
and this or that frequency? Sure you could  ….

Might a faster signal produce better results? I have a nasty
suspicion that the video was done with a signal that was
fast enough to show the best numbers possible on the ‘230.
That’s just a guess …..

Does the 53220 “surprise” also carry over to the 53210?
There’s no way to know. This test simply does not work
on a single channel counter. The 53180 is in the same
“twilight zone”.

How does a recently aligned and fully functional 5370
compare on a “fast edge” test. We have lots of data with
10 MHz…. Same question on the SR-620 and CNT-90.

Grubby details on the setup:

Both inputs AC coupled / 5V range
Auto Level turned off on both channels 
Channel A set to 1 M ohm
Channel B set to 50 ohms
Both trigger levels set to zero volts
Measure TI from A to B
My coax was about 2.5 ns long. 
External reference applied to all counters
External ref not coherent with the test signal. 

It took you longer to read all that nonsense than it will to
run the test :)

Fun !!!

Bob

Hi Back a few years ago, Keysight did one of their PR videos: https://www.youtube.com/watch?v=4DfLkxVe7Lk <https://www.youtube.com/watch?v=4DfLkxVe7Lk> It shows a fairly simple setup to see what your two channel counter is doing. A lot of counters have a built in standard deviation function so it applies to more than just the counter he’s showing. People have been doing this sort of test for many decades. The one thing he’s not to clear on is what sort of signal levels are being used. “Square wave” is about as specific as he gets. I decided to try the same test with a 100 MHz sine wave at 2.5 dbm. Sine wave is nice since it has a well defined slope at the zero crossing ( which does matter in this test). It also happened to be an available output on the FS-740. If you repeat the YouTube test with this signal, you get the same 13 ps one sigma and 9 ps resolution on a 53230A. That suggests to me that this is a “fast enough” signal to be useful. Same test gave the same results on a couple of 53230’s. So … what happens if we try this on a few other counters? A 53132A comes in right around 140 ps one sigma to give it a 100 ps resolution. That’s nice compared to a 150 ps spec. The “surprise” candidate is the 53220A. One would *think* that with a 100 ps resolution spec, it would be 5X worse than the 53230A. Not so much. It comes in at 14 ps one sigma and 10 ps resolution. ( Usual disclaimers apply …. used counter of unknown (China / eBay) origin …. one sample tested…. YMMV ….). Does this test tell you *everything* about your counter? No of course it doesn’t. There are a lot of bits and pieces in there, this is only looking at some of them. Does it tell you what you will see on a 10 MHz sine wave at 10 dbm? Nope, that’s a slower edge. The one sigma is roughly 24 ps on a 53230A. What it *does* provide is a quick check of how your counter is doing. Most of what gets killed in a counter is in the input channels (as noted in a previous post). This gives you a fast check on how they are doing. It also checks a bit more than that …. To be really useful, you need to run the test with *your* source on *your* counter(s). (It does not have to be 100 MHz and 2.5 dbm) Note the number(s) you get and file them away. Come back a year or three later scratching your head about some readings ….. you have something to check against. Checking against spec … not so much …. The key thing in this test as shown in the video *is* the fast edge. For a lot of years many of us have simply been using a 10 MHz test signal. There’s nothing wrong with doing that. One might *guess* that’s what the designers of the 53230 did. Turns out that a faster edge is a better edge, even at RF frequencies. Who knew …. ( other than anybody who watches YouTube …. :) ) Could you extend this test by playing with various levels to come up with a good idea of what to expect at this or that level and this or that frequency? Sure you could …. Might a faster signal produce better results? I have a nasty suspicion that the video was done with a signal that was fast enough to show the best numbers possible on the ‘230. That’s just a guess ….. Does the 53220 “surprise” also carry over to the 53210? There’s no way to know. This test simply does not work on a single channel counter. The 53180 is in the same “twilight zone”. How does a recently aligned and fully functional 5370 compare on a “fast edge” test. We have lots of data with 10 MHz…. Same question on the SR-620 and CNT-90. Grubby details on the setup: Both inputs AC coupled / 5V range Auto Level turned off on both channels Channel A set to 1 M ohm Channel B set to 50 ohms Both trigger levels set to zero volts Measure TI from A to B My coax was about 2.5 ns long. External reference applied to all counters External ref not coherent with the test signal. It took you longer to read all that nonsense than it will to run the test :) Fun !!! Bob
MD
Magnus Danielson
Wed, Feb 10, 2021 11:20 PM

Hi,

So, a very simple additional test one can do is to insert a 6 dB damping
pad. That effectively reduces the amplitude and also slashed the
slew-rate in half. You can do that for both squarewave and sine with the
same result. Using that, one can get some idea of the noise on the input
as you get two different slew-rates it converts to noise, and then see
how much "inherent" noise the counter has. Test with yet another
amplitude/slew-rate and you can refine that estimate and validate it.

If you want to have fun, try different trigger directions etc. to see
how well calibrated the inputs are.

Cheers,
Magnus

On 2021-02-10 18:18, Bob kb8tq wrote:

Hi

Back a few years ago, Keysight did one of their PR videos:

https://www.youtube.com/watch?v=4DfLkxVe7Lk https://www.youtube.com/watch?v=4DfLkxVe7Lk

It shows a fairly simple setup to see what your two channel
counter is doing. A lot of counters have a built in standard
deviation function so it applies to more than just the counter
he’s showing. People have been doing this sort of test for many
decades.

The one thing he’s not to clear on is what sort of signal levels
are being used. “Square wave” is about as specific as he gets.

I decided to try the same test with a 100 MHz sine wave at
2.5 dbm. Sine wave is nice since it has a well defined slope
at the zero crossing ( which does matter in this test). It also
happened to be an available output on the FS-740.

If you repeat the YouTube test with this signal, you get the same
13 ps one sigma and 9 ps resolution on a 53230A. That suggests
to me that this is a “fast enough” signal to be useful. Same test
gave the same results on a couple of 53230’s.

So … what happens if we try this on a few other counters?

A 53132A comes in right around 140 ps one sigma to give it
a 100 ps resolution. That’s nice compared to a 150 ps spec.

The “surprise” candidate is the 53220A. One would think that
with a 100 ps resolution spec, it would be 5X worse than the
53230A. Not so much. It comes in at 14 ps one sigma and 10 ps
resolution. ( Usual disclaimers apply …. used counter of
unknown (China / eBay) origin  …. one sample tested…. YMMV ….).

Does this test tell you everything about your counter? No
of course it doesn’t. There are a lot of bits and pieces in there,
this is only looking at some of them. Does it tell you what you
will see on a 10 MHz sine wave at 10 dbm? Nope, that’s a
slower edge. The one sigma is roughly 24 ps on a 53230A.

What it does provide is a quick check of how your counter
is doing. Most of what gets killed in a counter is in the input
channels (as noted in a previous post). This gives you a fast
check on how they are doing. It also checks a bit more than
that ….

To be really useful, you need to run the test with your source
on your counter(s). (It does not have to be 100 MHz and 2.5
dbm) Note the number(s) you get and file them away. Come back
a year or three later scratching your head about some readings
….. you have something to check against. Checking against spec
… not so much ….

The key thing in this test as shown in the video is the fast edge.
For a lot of years many of us have simply been using a 10 MHz
test signal. There’s nothing wrong with doing that. One might
guess that’s what the designers of the 53230 did. Turns out
that a faster edge is a better edge, even at RF frequencies.
Who knew …. ( other than anybody who watches YouTube …. :) )

Could you extend this test by playing with various levels to come
up with a good idea of what to expect at this or that level
and this or that frequency? Sure you could  ….

Might a faster signal produce better results? I have a nasty
suspicion that the video was done with a signal that was
fast enough to show the best numbers possible on the ‘230.
That’s just a guess …..

Does the 53220 “surprise” also carry over to the 53210?
There’s no way to know. This test simply does not work
on a single channel counter. The 53180 is in the same
“twilight zone”.

How does a recently aligned and fully functional 5370
compare on a “fast edge” test. We have lots of data with
10 MHz…. Same question on the SR-620 and CNT-90.

Grubby details on the setup:

Both inputs AC coupled / 5V range
Auto Level turned off on both channels 
Channel A set to 1 M ohm
Channel B set to 50 ohms
Both trigger levels set to zero volts
Measure TI from A to B
My coax was about 2.5 ns long. 
External reference applied to all counters
External ref not coherent with the test signal. 

It took you longer to read all that nonsense than it will to
run the test :)

Fun !!!

Bob


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi, So, a very simple additional test one can do is to insert a 6 dB damping pad. That effectively reduces the amplitude and also slashed the slew-rate in half. You can do that for both squarewave and sine with the same result. Using that, one can get some idea of the noise on the input as you get two different slew-rates it converts to noise, and then see how much "inherent" noise the counter has. Test with yet another amplitude/slew-rate and you can refine that estimate and validate it. If you want to have fun, try different trigger directions etc. to see how well calibrated the inputs are. Cheers, Magnus On 2021-02-10 18:18, Bob kb8tq wrote: > Hi > > Back a few years ago, Keysight did one of their PR videos: > > https://www.youtube.com/watch?v=4DfLkxVe7Lk <https://www.youtube.com/watch?v=4DfLkxVe7Lk> > > It shows a fairly simple setup to see what your two channel > counter is doing. A lot of counters have a built in standard > deviation function so it applies to more than just the counter > he’s showing. People have been doing this sort of test for many > decades. > > The one thing he’s not to clear on is what sort of signal levels > are being used. “Square wave” is about as specific as he gets. > > I decided to try the same test with a 100 MHz sine wave at > 2.5 dbm. Sine wave is nice since it has a well defined slope > at the zero crossing ( which does matter in this test). It also > happened to be an available output on the FS-740. > > If you repeat the YouTube test with this signal, you get the same > 13 ps one sigma and 9 ps resolution on a 53230A. That suggests > to me that this is a “fast enough” signal to be useful. Same test > gave the same results on a couple of 53230’s. > > So … what happens if we try this on a few other counters? > > A 53132A comes in right around 140 ps one sigma to give it > a 100 ps resolution. That’s nice compared to a 150 ps spec. > > The “surprise” candidate is the 53220A. One would *think* that > with a 100 ps resolution spec, it would be 5X worse than the > 53230A. Not so much. It comes in at 14 ps one sigma and 10 ps > resolution. ( Usual disclaimers apply …. used counter of > unknown (China / eBay) origin …. one sample tested…. YMMV ….). > > Does this test tell you *everything* about your counter? No > of course it doesn’t. There are a lot of bits and pieces in there, > this is only looking at some of them. Does it tell you what you > will see on a 10 MHz sine wave at 10 dbm? Nope, that’s a > slower edge. The one sigma is roughly 24 ps on a 53230A. > > What it *does* provide is a quick check of how your counter > is doing. Most of what gets killed in a counter is in the input > channels (as noted in a previous post). This gives you a fast > check on how they are doing. It also checks a bit more than > that …. > > To be really useful, you need to run the test with *your* source > on *your* counter(s). (It does not have to be 100 MHz and 2.5 > dbm) Note the number(s) you get and file them away. Come back > a year or three later scratching your head about some readings > ….. you have something to check against. Checking against spec > … not so much …. > > The key thing in this test as shown in the video *is* the fast edge. > For a lot of years many of us have simply been using a 10 MHz > test signal. There’s nothing wrong with doing that. One might > *guess* that’s what the designers of the 53230 did. Turns out > that a faster edge is a better edge, even at RF frequencies. > Who knew …. ( other than anybody who watches YouTube …. :) ) > > Could you extend this test by playing with various levels to come > up with a good idea of what to expect at this or that level > and this or that frequency? Sure you could …. > > Might a faster signal produce better results? I have a nasty > suspicion that the video was done with a signal that was > fast enough to show the best numbers possible on the ‘230. > That’s just a guess ….. > > Does the 53220 “surprise” also carry over to the 53210? > There’s no way to know. This test simply does not work > on a single channel counter. The 53180 is in the same > “twilight zone”. > > How does a recently aligned and fully functional 5370 > compare on a “fast edge” test. We have lots of data with > 10 MHz…. Same question on the SR-620 and CNT-90. > > Grubby details on the setup: > > Both inputs AC coupled / 5V range > Auto Level turned off on both channels > Channel A set to 1 M ohm > Channel B set to 50 ohms > Both trigger levels set to zero volts > Measure TI from A to B > My coax was about 2.5 ns long. > External reference applied to all counters > External ref not coherent with the test signal. > > It took you longer to read all that nonsense than it will to > run the test :) > > Fun !!! > > Bob > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
BK
Bob kb8tq
Wed, Feb 10, 2021 11:42 PM

Hi

I tried the pad on the 53230A. By the time I got to 6 db, the one sigma started
to come up. At 4 db … not so bad.

Bob

On Feb 10, 2021, at 6:20 PM, Magnus Danielson magnus@rubidium.se wrote:

Hi,

So, a very simple additional test one can do is to insert a 6 dB damping
pad. That effectively reduces the amplitude and also slashed the
slew-rate in half. You can do that for both squarewave and sine with the
same result. Using that, one can get some idea of the noise on the input
as you get two different slew-rates it converts to noise, and then see
how much "inherent" noise the counter has. Test with yet another
amplitude/slew-rate and you can refine that estimate and validate it.

If you want to have fun, try different trigger directions etc. to see
how well calibrated the inputs are.

Cheers,
Magnus

On 2021-02-10 18:18, Bob kb8tq wrote:

Hi

Back a few years ago, Keysight did one of their PR videos:

https://www.youtube.com/watch?v=4DfLkxVe7Lk https://www.youtube.com/watch?v=4DfLkxVe7Lk

It shows a fairly simple setup to see what your two channel
counter is doing. A lot of counters have a built in standard
deviation function so it applies to more than just the counter
he’s showing. People have been doing this sort of test for many
decades.

The one thing he’s not to clear on is what sort of signal levels
are being used. “Square wave” is about as specific as he gets.

I decided to try the same test with a 100 MHz sine wave at
2.5 dbm. Sine wave is nice since it has a well defined slope
at the zero crossing ( which does matter in this test). It also
happened to be an available output on the FS-740.

If you repeat the YouTube test with this signal, you get the same
13 ps one sigma and 9 ps resolution on a 53230A. That suggests
to me that this is a “fast enough” signal to be useful. Same test
gave the same results on a couple of 53230’s.

So … what happens if we try this on a few other counters?

A 53132A comes in right around 140 ps one sigma to give it
a 100 ps resolution. That’s nice compared to a 150 ps spec.

The “surprise” candidate is the 53220A. One would think that
with a 100 ps resolution spec, it would be 5X worse than the
53230A. Not so much. It comes in at 14 ps one sigma and 10 ps
resolution. ( Usual disclaimers apply …. used counter of
unknown (China / eBay) origin  …. one sample tested…. YMMV ….).

Does this test tell you everything about your counter? No
of course it doesn’t. There are a lot of bits and pieces in there,
this is only looking at some of them. Does it tell you what you
will see on a 10 MHz sine wave at 10 dbm? Nope, that’s a
slower edge. The one sigma is roughly 24 ps on a 53230A.

What it does provide is a quick check of how your counter
is doing. Most of what gets killed in a counter is in the input
channels (as noted in a previous post). This gives you a fast
check on how they are doing. It also checks a bit more than
that ….

To be really useful, you need to run the test with your source
on your counter(s). (It does not have to be 100 MHz and 2.5
dbm) Note the number(s) you get and file them away. Come back
a year or three later scratching your head about some readings
….. you have something to check against. Checking against spec
… not so much ….

The key thing in this test as shown in the video is the fast edge.
For a lot of years many of us have simply been using a 10 MHz
test signal. There’s nothing wrong with doing that. One might
guess that’s what the designers of the 53230 did. Turns out
that a faster edge is a better edge, even at RF frequencies.
Who knew …. ( other than anybody who watches YouTube …. :) )

Could you extend this test by playing with various levels to come
up with a good idea of what to expect at this or that level
and this or that frequency? Sure you could  ….

Might a faster signal produce better results? I have a nasty
suspicion that the video was done with a signal that was
fast enough to show the best numbers possible on the ‘230.
That’s just a guess …..

Does the 53220 “surprise” also carry over to the 53210?
There’s no way to know. This test simply does not work
on a single channel counter. The 53180 is in the same
“twilight zone”.

How does a recently aligned and fully functional 5370
compare on a “fast edge” test. We have lots of data with
10 MHz…. Same question on the SR-620 and CNT-90.

Grubby details on the setup:

Both inputs AC coupled / 5V range
Auto Level turned off on both channels 
Channel A set to 1 M ohm
Channel B set to 50 ohms
Both trigger levels set to zero volts
Measure TI from A to B
My coax was about 2.5 ns long. 
External reference applied to all counters
External ref not coherent with the test signal. 

It took you longer to read all that nonsense than it will to
run the test :)

Fun !!!

Bob


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi I tried the pad on the 53230A. By the time I got to 6 db, the one sigma started to come up. At 4 db … not so bad. Bob > On Feb 10, 2021, at 6:20 PM, Magnus Danielson <magnus@rubidium.se> wrote: > > Hi, > > So, a very simple additional test one can do is to insert a 6 dB damping > pad. That effectively reduces the amplitude and also slashed the > slew-rate in half. You can do that for both squarewave and sine with the > same result. Using that, one can get some idea of the noise on the input > as you get two different slew-rates it converts to noise, and then see > how much "inherent" noise the counter has. Test with yet another > amplitude/slew-rate and you can refine that estimate and validate it. > > If you want to have fun, try different trigger directions etc. to see > how well calibrated the inputs are. > > Cheers, > Magnus > > On 2021-02-10 18:18, Bob kb8tq wrote: >> Hi >> >> Back a few years ago, Keysight did one of their PR videos: >> >> https://www.youtube.com/watch?v=4DfLkxVe7Lk <https://www.youtube.com/watch?v=4DfLkxVe7Lk> >> >> It shows a fairly simple setup to see what your two channel >> counter is doing. A lot of counters have a built in standard >> deviation function so it applies to more than just the counter >> he’s showing. People have been doing this sort of test for many >> decades. >> >> The one thing he’s not to clear on is what sort of signal levels >> are being used. “Square wave” is about as specific as he gets. >> >> I decided to try the same test with a 100 MHz sine wave at >> 2.5 dbm. Sine wave is nice since it has a well defined slope >> at the zero crossing ( which does matter in this test). It also >> happened to be an available output on the FS-740. >> >> If you repeat the YouTube test with this signal, you get the same >> 13 ps one sigma and 9 ps resolution on a 53230A. That suggests >> to me that this is a “fast enough” signal to be useful. Same test >> gave the same results on a couple of 53230’s. >> >> So … what happens if we try this on a few other counters? >> >> A 53132A comes in right around 140 ps one sigma to give it >> a 100 ps resolution. That’s nice compared to a 150 ps spec. >> >> The “surprise” candidate is the 53220A. One would *think* that >> with a 100 ps resolution spec, it would be 5X worse than the >> 53230A. Not so much. It comes in at 14 ps one sigma and 10 ps >> resolution. ( Usual disclaimers apply …. used counter of >> unknown (China / eBay) origin …. one sample tested…. YMMV ….). >> >> Does this test tell you *everything* about your counter? No >> of course it doesn’t. There are a lot of bits and pieces in there, >> this is only looking at some of them. Does it tell you what you >> will see on a 10 MHz sine wave at 10 dbm? Nope, that’s a >> slower edge. The one sigma is roughly 24 ps on a 53230A. >> >> What it *does* provide is a quick check of how your counter >> is doing. Most of what gets killed in a counter is in the input >> channels (as noted in a previous post). This gives you a fast >> check on how they are doing. It also checks a bit more than >> that …. >> >> To be really useful, you need to run the test with *your* source >> on *your* counter(s). (It does not have to be 100 MHz and 2.5 >> dbm) Note the number(s) you get and file them away. Come back >> a year or three later scratching your head about some readings >> ….. you have something to check against. Checking against spec >> … not so much …. >> >> The key thing in this test as shown in the video *is* the fast edge. >> For a lot of years many of us have simply been using a 10 MHz >> test signal. There’s nothing wrong with doing that. One might >> *guess* that’s what the designers of the 53230 did. Turns out >> that a faster edge is a better edge, even at RF frequencies. >> Who knew …. ( other than anybody who watches YouTube …. :) ) >> >> Could you extend this test by playing with various levels to come >> up with a good idea of what to expect at this or that level >> and this or that frequency? Sure you could …. >> >> Might a faster signal produce better results? I have a nasty >> suspicion that the video was done with a signal that was >> fast enough to show the best numbers possible on the ‘230. >> That’s just a guess ….. >> >> Does the 53220 “surprise” also carry over to the 53210? >> There’s no way to know. This test simply does not work >> on a single channel counter. The 53180 is in the same >> “twilight zone”. >> >> How does a recently aligned and fully functional 5370 >> compare on a “fast edge” test. We have lots of data with >> 10 MHz…. Same question on the SR-620 and CNT-90. >> >> Grubby details on the setup: >> >> Both inputs AC coupled / 5V range >> Auto Level turned off on both channels >> Channel A set to 1 M ohm >> Channel B set to 50 ohms >> Both trigger levels set to zero volts >> Measure TI from A to B >> My coax was about 2.5 ns long. >> External reference applied to all counters >> External ref not coherent with the test signal. >> >> It took you longer to read all that nonsense than it will to >> run the test :) >> >> Fun !!! >> >> Bob >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.