Hi

With regard to the Transfer Accuracy spec for the 3458A, I have generally interpreted this as a measure of the differential linearity with a contribution from the short-term stability of the gain. In effect it becomes a measure of how reproducible a similar measurement could be over a short time. Hence it defines how well a value can be transferred from one device (say a Fluke 732A) to another (like a Fluke 5700A) at a similar voltage level (like 10V) over a short interval (like 10 minutes). For similar voltages this transfer doesn't really depend on the gain accuracy of the DVM,  only the gain stability.

Murray


This electronic transmission and any documents accompanying this electronic transmission contain confidential information belonging to the sender. This information may be legally privileged. The information is intended only for the use of the individual or entity named above. If you are not the intended recipient, you are hereby notified that any disclosure, copying, distribution or the taking of any action in reliance on or regarding the contents of this electronically transmitted information is strictly prohibited.

Hi With regard to the Transfer Accuracy spec for the 3458A, I have generally interpreted this as a measure of the differential linearity with a contribution from the short-term stability of the gain. In effect it becomes a measure of how reproducible a similar measurement could be over a short time. Hence it defines how well a value can be transferred from one device (say a Fluke 732A) to another (like a Fluke 5700A) at a similar voltage level (like 10V) over a short interval (like 10 minutes). For similar voltages this transfer doesn't really depend on the gain accuracy of the DVM, only the gain stability. Murray ________________________________ This electronic transmission and any documents accompanying this electronic transmission contain confidential information belonging to the sender. This information may be legally privileged. The information is intended only for the use of the individual or entity named above. If you are not the intended recipient, you are hereby notified that any disclosure, copying, distribution or the taking of any action in reliance on or regarding the contents of this electronically transmitted information is strictly prohibited.