Re: [volt-nuts] Fluke 5205a Repair

DG
David Garrido
Tue, Sep 29, 2015 6:23 PM

Hello All,

I am making a little more headway on this issue of the OVERLOAD & FAULT led’s being triggered and shutting down the system.

In paragraph 3-34 on page 3-4 of the manual it states:

Once U2-3 has been pulled low and a 3-9ms delay, the Operate FF J Input, U41-4 is supposed to go high.  And it does, BUT the next thing in the line is a short delay caused by R121 & C43 and that signal showing up at U41-1 should be low.  It is NOT and neither is the Q Output (U41-14).  So I thought ok GREAT, only pin 4 should be high and both of the others must be low.  Easy fix, plop in another SN74106N and I should be GTG.

Not so much………..still the same issue after warm-up, the OVERLOAD led comes on, followed very shortly by the FAULT led.  Now I checked U3-9, it is low and U3-8 high.  Next was U4-9 and it was high and pin 8 was low.  On to U40-9, it is low and pin 8 is high which is what was pushing U41-1 high.

So I went and checked R121, she measured 389 ohms and C43, a 4.7uf TA, measured 5.1 with 1.2 ESR.  That all seems good.

BUT……………….

So I decided to try putting the amp into the OPERATE mode and see if that changed the input to U41-1as it is supposed to follow the low from U3-8, it did.  Yes, U3-8 is now 100mV along with U3-9.  U4-9 is low and U4-8 is high.  U40-9 is high and 8 is low.  Now the signal at U41-4 is high as before, The signal at U41-1 is low as expected, BUT………..the signal at U41-14 did not go low following the low at U41-1.  Hmmmmmmmmmmm

Scratching my head………..

In summary in the standby mode:

U3-9 LOW
U3-8 HIGH

U4-9 HIGH
U4-8 LOW

U40-9 LOW
U40-8 HIGH

U41-1 HIGH
U41-4 HIGH
U41-14 HIGH

In summary in the OPERATE mode:

U3-9 LOW
U3-8 LOW

U4-9 LOW
U4-8 HIGH

U40-9 HIGH
U40-8 LOW

U41-1 LOW
U41-4 HIGH
U41-14 HIGH

I could use a little more brain power as this too seems to be a bad U41-FF, but I changed that.  So unless my other U41 FF was bad in the very same way……………what am I missing????

Thoughts?

Cheers,

David

Hello All, I am making a little more headway on this issue of the OVERLOAD & FAULT led’s being triggered and shutting down the system. In paragraph 3-34 on page 3-4 of the manual it states: Once U2-3 has been pulled low and a 3-9ms delay, the Operate FF J Input, U41-4 is supposed to go high. And it does, BUT the next thing in the line is a short delay caused by R121 & C43 and that signal showing up at U41-1 should be low. It is NOT and neither is the Q Output (U41-14). So I thought ok GREAT, only pin 4 should be high and both of the others must be low. Easy fix, plop in another SN74106N and I should be GTG. Not so much………..still the same issue after warm-up, the OVERLOAD led comes on, followed very shortly by the FAULT led. Now I checked U3-9, it is low and U3-8 high. Next was U4-9 and it was high and pin 8 was low. On to U40-9, it is low and pin 8 is high which is what was pushing U41-1 high. So I went and checked R121, she measured 389 ohms and C43, a 4.7uf TA, measured 5.1 with 1.2 ESR. That all seems good. BUT………………. So I decided to try putting the amp into the OPERATE mode and see if that changed the input to U41-1as it is supposed to follow the low from U3-8, it did. Yes, U3-8 is now 100mV along with U3-9. U4-9 is low and U4-8 is high. U40-9 is high and 8 is low. Now the signal at U41-4 is high as before, The signal at U41-1 is low as expected, BUT………..the signal at U41-14 did not go low following the low at U41-1. Hmmmmmmmmmmm Scratching my head……….. In summary in the standby mode: U3-9 LOW U3-8 HIGH U4-9 HIGH U4-8 LOW U40-9 LOW U40-8 HIGH U41-1 HIGH U41-4 HIGH U41-14 HIGH In summary in the OPERATE mode: U3-9 LOW U3-8 LOW U4-9 LOW U4-8 HIGH U40-9 HIGH U40-8 LOW U41-1 LOW U41-4 HIGH U41-14 HIGH I could use a little more brain power as this too seems to be a bad U41-FF, but I changed that. So unless my other U41 FF was bad in the very same way……………what am I missing???? Thoughts? Cheers, David
CH
Chuck Harris
Tue, Sep 29, 2015 8:06 PM

Hi David,

My eye keeps wandering over to U40, U43, and that abortion of
tantalum capacitors, and snubber networks hung on the inputs to
U40, U1 and U2.

Whenever I see stuff like that, I know the engineer was getting
frantic over being toasted by glitches.  FF clocks just aren't
at all amused when they see a glitch come on their clock, or
preset/reset lines.

Also, old tantalums have a habit of becoming noise sources as
they short, and unshort bits and pieces of their oxide layer.

I have seen a few that tested ok, except that their leakage current
kept spiking and returning to normal. You might not notice that
sort of behavior on a power supply bypass, but it is bound to
raise an eyebrow when it happens on a logic input line, or
an integrator (not this circuit, I know).

Capacitors hung on logic lines are always a bandaid fix.

-Chuck Harris

David Garrido wrote:

Hello All,

I am making a little more headway on this issue of the OVERLOAD & FAULT led’s
being triggered and shutting down the system.

In paragraph 3-34 on page 3-4 of the manual it states:

Once U2-3 has been pulled low and a 3-9ms delay, the Operate FF J Input, U41-4 is
supposed to go high.  And it does, BUT the next thing in the line is a short delay
caused by R121 & C43 and that signal showing up at U41-1 should be low.  It is NOT
and neither is the Q Output (U41-14).  So I thought ok GREAT, only pin 4 should be
high and both of the others must be low.  Easy fix, plop in another SN74106N and I
should be GTG.

Not so much………..still the same issue after warm-up, the OVERLOAD led comes on,
followed very shortly by the FAULT led.  Now I checked U3-9, it is low and U3-8
high.  Next was U4-9 and it was high and pin 8 was low.  On to U40-9, it is low
and pin 8 is high which is what was pushing U41-1 high.

So I went and checked R121, she measured 389 ohms and C43, a 4.7uf TA, measured
5.1 with 1.2 ESR.  That all seems good.

BUT……………….

So I decided to try putting the amp into the OPERATE mode and see if that changed
the input to U41-1as it is supposed to follow the low from U3-8, it did.  Yes,
U3-8 is now 100mV along with U3-9.  U4-9 is low and U4-8 is high.  U40-9 is high
and 8 is low.  Now the signal at U41-4 is high as before, The signal at U41-1 is
low as expected, BUT………..the signal at U41-14 did not go low following the low at
U41-1.  Hmmmmmmmmmmm

Scratching my head………..

In summary in the standby mode:

U3-9 LOW U3-8 HIGH

U4-9 HIGH U4-8 LOW

U40-9 LOW U40-8 HIGH

U41-1 HIGH U41-4 HIGH U41-14 HIGH

In summary in the OPERATE mode:

U3-9 LOW U3-8 LOW

U4-9 LOW U4-8 HIGH

U40-9 HIGH U40-8 LOW

U41-1 LOW U41-4 HIGH U41-14 HIGH

I could use a little more brain power as this too seems to be a bad U41-FF, but I
changed that.  So unless my other U41 FF was bad in the very same way……………what am
I missing????

Thoughts?

Cheers,

David

_______________________________________________ volt-nuts mailing list --
volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the
instructions there.

Hi David, My eye keeps wandering over to U40, U43, and that abortion of tantalum capacitors, and snubber networks hung on the inputs to U40, U1 and U2. Whenever I see stuff like that, I know the engineer was getting frantic over being toasted by glitches. FF clocks just aren't at all amused when they see a glitch come on their clock, or preset/reset lines. Also, old tantalums have a habit of becoming noise sources as they short, and unshort bits and pieces of their oxide layer. I have seen a few that tested ok, except that their leakage current kept spiking and returning to normal. You might not notice that sort of behavior on a power supply bypass, but it is bound to raise an eyebrow when it happens on a logic input line, or an integrator (not this circuit, I know). Capacitors hung on logic lines are always a bandaid fix. -Chuck Harris David Garrido wrote: > Hello All, > > I am making a little more headway on this issue of the OVERLOAD & FAULT led’s > being triggered and shutting down the system. > > In paragraph 3-34 on page 3-4 of the manual it states: > > Once U2-3 has been pulled low and a 3-9ms delay, the Operate FF J Input, U41-4 is > supposed to go high. And it does, BUT the next thing in the line is a short delay > caused by R121 & C43 and that signal showing up at U41-1 should be low. It is NOT > and neither is the Q Output (U41-14). So I thought ok GREAT, only pin 4 should be > high and both of the others must be low. Easy fix, plop in another SN74106N and I > should be GTG. > > Not so much………..still the same issue after warm-up, the OVERLOAD led comes on, > followed very shortly by the FAULT led. Now I checked U3-9, it is low and U3-8 > high. Next was U4-9 and it was high and pin 8 was low. On to U40-9, it is low > and pin 8 is high which is what was pushing U41-1 high. > > So I went and checked R121, she measured 389 ohms and C43, a 4.7uf TA, measured > 5.1 with 1.2 ESR. That all seems good. > > BUT………………. > > So I decided to try putting the amp into the OPERATE mode and see if that changed > the input to U41-1as it is supposed to follow the low from U3-8, it did. Yes, > U3-8 is now 100mV along with U3-9. U4-9 is low and U4-8 is high. U40-9 is high > and 8 is low. Now the signal at U41-4 is high as before, The signal at U41-1 is > low as expected, BUT………..the signal at U41-14 did not go low following the low at > U41-1. Hmmmmmmmmmmm > > Scratching my head……….. > > In summary in the standby mode: > > U3-9 LOW U3-8 HIGH > > U4-9 HIGH U4-8 LOW > > U40-9 LOW U40-8 HIGH > > U41-1 HIGH U41-4 HIGH U41-14 HIGH > > In summary in the OPERATE mode: > > U3-9 LOW U3-8 LOW > > U4-9 LOW U4-8 HIGH > > U40-9 HIGH U40-8 LOW > > U41-1 LOW U41-4 HIGH U41-14 HIGH > > > I could use a little more brain power as this too seems to be a bad U41-FF, but I > changed that. So unless my other U41 FF was bad in the very same way……………what am > I missing???? > > Thoughts? > > Cheers, > > David > > > _______________________________________________ volt-nuts mailing list -- > volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the > instructions there. >