time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

HP 5370B low frequency modulation

BG
Björn Gabrielsson
Mon, Jul 23, 2007 7:34 AM

Hej Magnus,

Did you consider switching the 5370 and the 5372 oscillators? Putting the
blame either with the osc or the hosting 5370.

--

Björn

On Mon, July 23, 2007 5:10, Magnus Danielson said:

); SAEximRunCond expanded to false
Errors-To: time-nuts-bounces+bg=lysator.liu.se+bg=lysator.liu.se@febo.com

From: "Mike Feher" mfeher@eozinc.com
Subject: Re: [time-nuts] HP 5370B low frequency modulation
Date: Sun, 22 Jul 2007 22:58:00 -0400
Message-ID: 00b101c7ccd5$490ae6a0$0201a8c0@gsmacdq14es

Mike,

Highly unlikely, but, possible, especially if it was in a corrosive
atmosphere. Of course then I would expect to see evidence of corrosion
on
other components. When we do not know the answer we can come up with so
many
possibilities and not have one of them correct. I have run across a lot
of
those defective fuses over the years as HP has been using them forever.
There was no real evidence as to why they opened, and that is why before
I
guessed at just plain ageing and the wax eventually getting soft. I just
checked the current flow within the oscillator where the fuse used to be
plugged in, which was at least nice of them, and always found it to be
normal. I just took a piece of buss wire and plugged it into the same
two
pins were the fuse used to be and never had a problem again. The
situation
Magnus describes of course is very unusual, and, in reality would only
be
found by a time nut as it is so miniscule. There is also some
periodicity to
it, which does suggest some control loop problem, but, I would hate to
even
take a guess. - Mike

May I also point out that there may be as low as short low period per 1000
s or
about 3-8 (rought estimate).

I am considering all kinds of possible reasons for it. The group certainly
has
more experience shared among them than I have on these, so I concentrate
on
observations.

I wonder if it may be some form of initial shock burnout that I am
witnessing.
I have no idea how they are suppposed to look. It is not like you want to
toss
your 10811s to the floor just to see how they behave as a result, now is
there?
If I where making them I would, but with a thad more of science attached
to it.

Cheers,
Magnus


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hej Magnus, Did you consider switching the 5370 and the 5372 oscillators? Putting the blame either with the osc or the hosting 5370. -- Björn On Mon, July 23, 2007 5:10, Magnus Danielson said: > ); SAEximRunCond expanded to false > Errors-To: time-nuts-bounces+bg=lysator.liu.se+bg=lysator.liu.se@febo.com > > From: "Mike Feher" <mfeher@eozinc.com> > Subject: Re: [time-nuts] HP 5370B low frequency modulation > Date: Sun, 22 Jul 2007 22:58:00 -0400 > Message-ID: <00b101c7ccd5$490ae6a0$0201a8c0@gsmacdq14es> > > Mike, > >> Highly unlikely, but, possible, especially if it was in a corrosive >> atmosphere. Of course then I would expect to see evidence of corrosion >> on >> other components. When we do not know the answer we can come up with so >> many >> possibilities and not have one of them correct. I have run across a lot >> of >> those defective fuses over the years as HP has been using them forever. >> There was no real evidence as to why they opened, and that is why before >> I >> guessed at just plain ageing and the wax eventually getting soft. I just >> checked the current flow within the oscillator where the fuse used to be >> plugged in, which was at least nice of them, and always found it to be >> normal. I just took a piece of buss wire and plugged it into the same >> two >> pins were the fuse used to be and never had a problem again. The >> situation >> Magnus describes of course is very unusual, and, in reality would only >> be >> found by a time nut as it is so miniscule. There is also some >> periodicity to >> it, which does suggest some control loop problem, but, I would hate to >> even >> take a guess. - Mike > > May I also point out that there may be as low as short low period per 1000 > s or > about 3-8 (rought estimate). > > I am considering all kinds of possible reasons for it. The group certainly > has > more experience shared among them than I have on these, so I concentrate > on > observations. > > I wonder if it may be some form of initial shock burnout that I am > witnessing. > I have no idea how they are suppposed to look. It is not like you want to > toss > your 10811s to the floor just to see how they behave as a result, now is > there? > If I where making them I would, but with a thad more of science attached > to it. > > Cheers, > Magnus > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
MD
Magnus Danielson
Mon, Jul 23, 2007 9:48 AM

From: Björn_Gabrielsson bg@lysator.liu.se
Subject: Re: [time-nuts] HP 5370B low frequency modulation
Date: Mon, 23 Jul 2007 09:34:35 +0200 (CEST)
Message-ID: 62656.87.227.52.225.1185176075.squirrel@webmail.lysator.liu.se

); SAEximRunCond expanded to false
Errors-To: time-nuts-bounces+magnus=rubidium.dyndns.org+magnus=rubidium.dyndns.org@febo.com

Hej Magnus,

Hej Björn,

Did you consider switching the 5370 and the 5372 oscillators? Putting the
blame either with the osc or the hosting 5370.

No, good thinking. That would be possible but there is a bit of a logistical
problem prohibiting that as a quick-test. The 5372 sits under a Tek 547 scope
which weighs alot.

Cheers,
Magnus

From: Björn_Gabrielsson <bg@lysator.liu.se> Subject: Re: [time-nuts] HP 5370B low frequency modulation Date: Mon, 23 Jul 2007 09:34:35 +0200 (CEST) Message-ID: <62656.87.227.52.225.1185176075.squirrel@webmail.lysator.liu.se> > ); SAEximRunCond expanded to false > Errors-To: time-nuts-bounces+magnus=rubidium.dyndns.org+magnus=rubidium.dyndns.org@febo.com > > Hej Magnus, Hej Björn, > Did you consider switching the 5370 and the 5372 oscillators? Putting the > blame either with the osc or the hosting 5370. No, good thinking. That would be possible but there is a bit of a logistical problem prohibiting that as a quick-test. The 5372 sits under a Tek 547 scope which weighs alot. Cheers, Magnus
DJ
Didier Juges
Mon, Jul 23, 2007 10:26 AM

I have seen cold solder joints on thermal fuses and certain types of
capacitors, while the rest of the instrument was fine with no sign of
corrosion.

I think it has to do with the metal used for certain component leads. Either
they were never soldered well, or interface corrosion developed over time.

Gold plated connectors are a well known example.

Didier KO4BB

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Mike Feher
Sent: Sunday, July 22, 2007 9:58 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] HP 5370B low frequency modulation

Highly unlikely, but, possible, especially if it was in a corrosive
atmosphere. Of course then I would expect to see evidence of corrosion on
other components.
...

Mike B. Feher, N4FS
89 Arnold Blvd.
Howell, NJ, 07731
732-886-5960

I have seen cold solder joints on thermal fuses and certain types of capacitors, while the rest of the instrument was fine with no sign of corrosion. I think it has to do with the metal used for certain component leads. Either they were never soldered well, or interface corrosion developed over time. Gold plated connectors are a well known example. Didier KO4BB -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of Mike Feher Sent: Sunday, July 22, 2007 9:58 PM To: 'Discussion of precise time and frequency measurement' Subject: Re: [time-nuts] HP 5370B low frequency modulation Highly unlikely, but, possible, especially if it was in a corrosive atmosphere. Of course then I would expect to see evidence of corrosion on other components. ... Mike B. Feher, N4FS 89 Arnold Blvd. Howell, NJ, 07731 732-886-5960
BC
Brooke Clarke
Mon, Jul 23, 2007 1:53 PM

Hi Didier:

Would you elaborate on the comment "Gold plated connectors are a well known
example."  Do you mean when soldered with Lead Tin solder instead of a silver
bearing solder or something else?

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com

Didier Juges wrote:

); SAEximRunCond expanded to false
Errors-To: time-nuts-bounces+brooke=pacific.net+brooke=pacific.net@febo.com

I have seen cold solder joints on thermal fuses and certain types of
capacitors, while the rest of the instrument was fine with no sign of
corrosion.

I think it has to do with the metal used for certain component leads. Either
they were never soldered well, or interface corrosion developed over time.

Gold plated connectors are a well known example.

Didier KO4BB

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Mike Feher
Sent: Sunday, July 22, 2007 9:58 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] HP 5370B low frequency modulation

Highly unlikely, but, possible, especially if it was in a corrosive
atmosphere. Of course then I would expect to see evidence of corrosion on
other components.
...

Mike B. Feher, N4FS
89 Arnold Blvd.
Howell, NJ, 07731
732-886-5960


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Didier: Would you elaborate on the comment "Gold plated connectors are a well known example." Do you mean when soldered with Lead Tin solder instead of a silver bearing solder or something else? Have Fun, Brooke Clarke http://www.PRC68.com http://www.precisionclock.com Didier Juges wrote: > ); SAEximRunCond expanded to false > Errors-To: time-nuts-bounces+brooke=pacific.net+brooke=pacific.net@febo.com > > I have seen cold solder joints on thermal fuses and certain types of > capacitors, while the rest of the instrument was fine with no sign of > corrosion. > > I think it has to do with the metal used for certain component leads. Either > they were never soldered well, or interface corrosion developed over time. > > Gold plated connectors are a well known example. > > Didier KO4BB > > -----Original Message----- > From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On > Behalf Of Mike Feher > Sent: Sunday, July 22, 2007 9:58 PM > To: 'Discussion of precise time and frequency measurement' > Subject: Re: [time-nuts] HP 5370B low frequency modulation > > Highly unlikely, but, possible, especially if it was in a corrosive > atmosphere. Of course then I would expect to see evidence of corrosion on > other components. > ... > > Mike B. Feher, N4FS > 89 Arnold Blvd. > Howell, NJ, 07731 > 732-886-5960 > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
P
Pete
Mon, Jul 23, 2007 2:14 PM

Magnus,

Have you used your 5372A reference output to drive the 5370B ref in?
Certainly easier than switching 10811As.

Pete Rawson

Magnus, Have you used your 5372A reference output to drive the 5370B ref in? Certainly easier than switching 10811As. Pete Rawson
DB
Dr Bruce Griffiths
Mon, Jul 23, 2007 2:28 PM

Brooke Clarke wrote:

); SAEximRunCond expanded to false
Errors-To: time-nuts-bounces+bruce.griffiths=xtra.co.nz+bruce.griffiths=xtra.co.nz@febo.com

Hi Didier:

Would you elaborate on the comment "Gold plated connectors are a well known
example."  Do you mean when soldered with Lead Tin solder instead of a silver
bearing solder or something else?

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com

Didier Juges wrote:

); SAEximRunCond expanded to false
Errors-To: time-nuts-bounces+brooke=pacific.net+brooke=pacific.net@febo.com

I have seen cold solder joints on thermal fuses and certain types of
capacitors, while the rest of the instrument was fine with no sign of
corrosion.

I think it has to do with the metal used for certain component leads. Either
they were never soldered well, or interface corrosion developed over time.

Gold plated connectors are a well known example.

Didier KO4BB

Brooke

Gold dissolves in the solder and a gold -tin intermetallic compound is
formed which severely reduces the joint integrity and ductility.
This can be circumvented by keeping the gold concentration in the solder
below 4%. Tin plating over the gold before soldering is sometimes used
to ensure this.

Bruce

Brooke Clarke wrote: > ); SAEximRunCond expanded to false > Errors-To: time-nuts-bounces+bruce.griffiths=xtra.co.nz+bruce.griffiths=xtra.co.nz@febo.com > > Hi Didier: > > Would you elaborate on the comment "Gold plated connectors are a well known > example." Do you mean when soldered with Lead Tin solder instead of a silver > bearing solder or something else? > > Have Fun, > > Brooke Clarke > http://www.PRC68.com > http://www.precisionclock.com > > > > Didier Juges wrote: > >> ); SAEximRunCond expanded to false >> Errors-To: time-nuts-bounces+brooke=pacific.net+brooke=pacific.net@febo.com >> >> I have seen cold solder joints on thermal fuses and certain types of >> capacitors, while the rest of the instrument was fine with no sign of >> corrosion. >> >> I think it has to do with the metal used for certain component leads. Either >> they were never soldered well, or interface corrosion developed over time. >> >> Gold plated connectors are a well known example. >> >> Didier KO4BB >> Brooke Gold dissolves in the solder and a gold -tin intermetallic compound is formed which severely reduces the joint integrity and ductility. This can be circumvented by keeping the gold concentration in the solder below 4%. Tin plating over the gold before soldering is sometimes used to ensure this. Bruce
ML
Maggie Leber
Mon, Jul 23, 2007 3:05 PM

Maybe some of you haven't seen this oscilloscope clock display yet:

http://www.makezine.com/blog/archive/2007/07/avr_oscilloscope_clock.html

--
Margaret Stephanie Leber CCP, SCJP, SCWCD
http://voicenet.com/~maggie
AOPA 925383 - Amateur Radio Station K3XS - ARRL 39280 - AMSAT 32844
"The art of progress is to preserve order amid change
and to preserve change amid order."-A.N.Whitehead

Maybe some of you haven't seen this oscilloscope clock display yet: http://www.makezine.com/blog/archive/2007/07/avr_oscilloscope_clock.html -- Margaret Stephanie Leber CCP, SCJP, SCWCD http://voicenet.com/~maggie AOPA 925383 - Amateur Radio Station K3XS - ARRL 39280 - AMSAT 32844 "The art of progress is to preserve order amid change and to preserve change amid order."-A.N.Whitehead
MD
Magnus Danielson
Tue, Jul 24, 2007 1:50 AM

From: "Pete" peterawson@earthlink.net
Subject: Re: [time-nuts] HP 5370B low frequency modulation
Date: Mon, 23 Jul 2007 08:14:32 -0600
Message-ID: 002601c7cd33$ce56f520$0200a8c0@BASE1

); SAEximRunCond expanded to false
Errors-To: time-nuts-bounces+magnus=rubidium.dyndns.org+magnus=rubidium.dyndns.org@febo.com

Magnus,

Pete,

Have you used your 5372A reference output to drive the 5370B ref in?
Certainly easier than switching 10811As.

Lack of wiring isle (wrong spelling last time, sorry) prohibited that. It was
just too hard to get to at that moment. As I have another 10811 inside another
instrument I used that as a reference instead. Thanks to fellow time-nut Björn
who I just came home from I have an easier swap-option exercise to do as he was
kind enought to borrow me another 10811 which was easier to get to. Not in
metrical distance thought, but who cares, I was going there anyways. :-)

Anyway, I have no reason to think that the output buffer board would subtract
frequency for length periods of time, up to a minute. It would mean that what
is more or less amplifiers and logical gates would drop cycles with only one
dynamic and one static signal. It is much more reasnoble to assume that the
10811 itself experience some modulation.

Cheers,
Magnus

From: "Pete" <peterawson@earthlink.net> Subject: Re: [time-nuts] HP 5370B low frequency modulation Date: Mon, 23 Jul 2007 08:14:32 -0600 Message-ID: <002601c7cd33$ce56f520$0200a8c0@BASE1> > ); SAEximRunCond expanded to false > Errors-To: time-nuts-bounces+magnus=rubidium.dyndns.org+magnus=rubidium.dyndns.org@febo.com > > Magnus, Pete, > Have you used your 5372A reference output to drive the 5370B ref in? > Certainly easier than switching 10811As. Lack of wiring isle (wrong spelling last time, sorry) prohibited that. It was just too hard to get to at that moment. As I have another 10811 inside another instrument I used that as a reference instead. Thanks to fellow time-nut Björn who I just came home from I have an easier swap-option exercise to do as he was kind enought to borrow me another 10811 which was easier to get to. Not in metrical distance thought, but who cares, I was going there anyways. :-) Anyway, I have no reason to think that the output buffer board would subtract frequency for length periods of time, up to a minute. It would mean that what is more or less amplifiers and logical gates would drop cycles with only one dynamic and one static signal. It is much more reasnoble to assume that the 10811 itself experience some modulation. Cheers, Magnus
DJ
Didier Juges
Tue, Jul 24, 2007 2:12 AM

Brooke,

Thanks Bruce for the quick response, as always.

In my business, we typically dissolve the gold by dipping the solder cup in
a hot tin pot, which is replaced regularly, before soldering the wires. This
has been blessed by the major defense contractors for military applications.

Failure to do this typically results in weakening solder joints over time,
and delayed failures, particularly when there is mechanical and/or thermal
stress applied to the joint.

Didier KO4BB

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Dr Bruce Griffiths
Sent: Monday, July 23, 2007 9:29 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B low frequency modulation

Brooke Clarke wrote:

); SAEximRunCond expanded to false
Errors-To:

time-nuts-bounces+bruce.griffiths=xtra.co.nz+bruce.griffiths=xtra.co.nz@febo
.com

Hi Didier:

Would you elaborate on the comment "Gold plated connectors are a well

known

example."  Do you mean when soldered with Lead Tin solder instead of a

silver

bearing solder or something else?

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com

Didier Juges wrote:

); SAEximRunCond expanded to false
Errors-To:

time-nuts-bounces+brooke=pacific.net+brooke=pacific.net@febo.com

I have seen cold solder joints on thermal fuses and certain types of
capacitors, while the rest of the instrument was fine with no sign of
corrosion.

I think it has to do with the metal used for certain component leads.

Either

they were never soldered well, or interface corrosion developed over

time.

Gold plated connectors are a well known example.

Didier KO4BB

Brooke

Gold dissolves in the solder and a gold -tin intermetallic compound is
formed which severely reduces the joint integrity and ductility.
This can be circumvented by keeping the gold concentration in the solder
below 4%. Tin plating over the gold before soldering is sometimes used
to ensure this.

Bruce


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Brooke, Thanks Bruce for the quick response, as always. In my business, we typically dissolve the gold by dipping the solder cup in a hot tin pot, which is replaced regularly, before soldering the wires. This has been blessed by the major defense contractors for military applications. Failure to do this typically results in weakening solder joints over time, and delayed failures, particularly when there is mechanical and/or thermal stress applied to the joint. Didier KO4BB -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of Dr Bruce Griffiths Sent: Monday, July 23, 2007 9:29 AM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] HP 5370B low frequency modulation Brooke Clarke wrote: > ); SAEximRunCond expanded to false > Errors-To: time-nuts-bounces+bruce.griffiths=xtra.co.nz+bruce.griffiths=xtra.co.nz@febo .com > > Hi Didier: > > Would you elaborate on the comment "Gold plated connectors are a well known > example." Do you mean when soldered with Lead Tin solder instead of a silver > bearing solder or something else? > > Have Fun, > > Brooke Clarke > http://www.PRC68.com > http://www.precisionclock.com > > > > Didier Juges wrote: > >> ); SAEximRunCond expanded to false >> Errors-To: time-nuts-bounces+brooke=pacific.net+brooke=pacific.net@febo.com >> >> I have seen cold solder joints on thermal fuses and certain types of >> capacitors, while the rest of the instrument was fine with no sign of >> corrosion. >> >> I think it has to do with the metal used for certain component leads. Either >> they were never soldered well, or interface corrosion developed over time. >> >> Gold plated connectors are a well known example. >> >> Didier KO4BB >> Brooke Gold dissolves in the solder and a gold -tin intermetallic compound is formed which severely reduces the joint integrity and ductility. This can be circumvented by keeping the gold concentration in the solder below 4%. Tin plating over the gold before soldering is sometimes used to ensure this. Bruce _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
MD
Magnus Danielson
Tue, Jul 24, 2007 2:36 AM

From: "Didier Juges" didier@cox.net
Subject: Re: [time-nuts] HP 5370B low frequency modulation
Date: Mon, 23 Jul 2007 21:12:39 -0500
Message-ID: 000001c7cd98$1c8d3dd0$6601a8c0@didierhp

Brooke,

Thanks Bruce for the quick response, as always.

In my business, we typically dissolve the gold by dipping the solder cup in
a hot tin pot, which is replaced regularly, before soldering the wires. This
has been blessed by the major defense contractors for military applications.

Failure to do this typically results in weakening solder joints over time,
and delayed failures, particularly when there is mechanical and/or thermal
stress applied to the joint.

Gold will dissolve very quickly into the Sn/Pb solder.

If you have a too high gold content in the joint, it can form one of several
Au/Sn mixtures which is brittle and will crack easilly. We had trouble with
that under our BGAs so when the gold layer was too thick, the whole 400-500
ball BGA could crack loose of the board if you accidentilly hit the rather
flimsy cooling flange. On another board we glued a nut to the top of the BGA
and hanged 50-70 kg of it and while the board was clearly bent the BGA was
still firmly hooked to it. Now that's the difference between a few um of gold
too much or not.

There are other cases where the gold-plating is not an issue.

This has been covered in public sources, so a dig around would give some
indication.

Cheers,
Magnus

From: "Didier Juges" <didier@cox.net> Subject: Re: [time-nuts] HP 5370B low frequency modulation Date: Mon, 23 Jul 2007 21:12:39 -0500 Message-ID: <000001c7cd98$1c8d3dd0$6601a8c0@didierhp> > Brooke, > > Thanks Bruce for the quick response, as always. > > In my business, we typically dissolve the gold by dipping the solder cup in > a hot tin pot, which is replaced regularly, before soldering the wires. This > has been blessed by the major defense contractors for military applications. > > Failure to do this typically results in weakening solder joints over time, > and delayed failures, particularly when there is mechanical and/or thermal > stress applied to the joint. Gold will dissolve very quickly into the Sn/Pb solder. If you have a too high gold content in the joint, it can form one of several Au/Sn mixtures which is brittle and will crack easilly. We had trouble with that under our BGAs so when the gold layer was too thick, the whole 400-500 ball BGA could crack loose of the board if you accidentilly hit the rather flimsy cooling flange. On another board we glued a nut to the top of the BGA and hanged 50-70 kg of it and while the board was clearly bent the BGA was still firmly hooked to it. Now that's the difference between a few um of gold too much or not. There are other cases where the gold-plating is not an issue. This has been covered in public sources, so a dig around would give some indication. Cheers, Magnus