volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Fluke 731B battery & charge circuit

OE
Orin Eman
Sun, Jun 23, 2013 8:27 PM

Well, an E505 and some 1n967Bs showed up yesterday.  I breadboarded them
with a 180 ohm resistor and a bench power supply that goes to 30V.

The 1N967Bs ranged from 16.8V to 18.1V when driven with the E505.  Voltage
across the resistor was about 0.195, so about 1.1mA.

I chose the 18.1V zener and started the fun and games of installing them.
Someone had soldered the mains connectors.  Hard to see what was going on
under the heat shrink.  Hard to apply enough heat to get them off without
lifting pads from the board.  I'm going to have to do some repair here -
find some new pins/connectors.  It also looks like a previous battery has
leaked and there is some corrosion in places.

The part that had been used to replace the E505 in the past is marked
2789.  Anyone recognize that?

This was a disaster.  Although fine on battery, and even with the battery
disconnected, on AC, output voltage is millivolts(!) low.  So I connect up
my 3455A to the pre-supply output and the output is now correct.  Poked
around with the 'scope and same problem, as soon as I connect it, the
problem goes away.  I figure something is oscillating and connecting
anything damps it.  I can see the problem on the 10V output with the
scope.  2mV jumps up and down.  Not a square wave, but perhaps 25% duty
cycle at 120Hz.  I did eventually manage to catch a similar waveform at the
emitter of the pass transistor.  It would vary in amplitude - up to 40mV
peak, dropping if I put my hand near the probe.

After a couple of hours tinkering (try a zener that was 17.5V, capacitor
across the zener etc.), I gave up and put the '2789' and original zener/2
1N4148s back in.  The 10V output is back to normal, but the 1.018 setting
is reading 1.0V.  A careful inspection found that the grey wire had broken
off the switch - a relatively easy fix.

What next?  Perhaps try the 18.1V zener - at least I can replace that from
the top of the board.  I don't feel like taking everything apart again to
swap the '2789' out again.

Orin.

On Thu, Jun 20, 2013 at 1:49 PM, Bob Smither smither@c-c-i.com wrote:

On 06/17/2013 02:29 AM, Christopher Brown wrote:

<snip>

That jfet and the shielded transformer are hard to get.  Everything else
on the power supply board is nothing special.  All of the specially
selected stuff is on the daughter board.

<snip>

I needed the current diode to repair a 731B as well.  Try EBay.com.
Search for
"j505 current" and you will find them.

--


---=======
Bob Smither, PhD                                  Circuit Concepts, Inc.

"There is no worse tyranny than to force a man to pay for what
he does not want merely because you think it would be good for him."
-- Robert Heinlein

Smither@C-C-I.Com            http://www.C-C-I.Com            281-331-2744


---=======


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Well, an E505 and some 1n967Bs showed up yesterday. I breadboarded them with a 180 ohm resistor and a bench power supply that goes to 30V. The 1N967Bs ranged from 16.8V to 18.1V when driven with the E505. Voltage across the resistor was about 0.195, so about 1.1mA. I chose the 18.1V zener and started the fun and games of installing them. Someone had soldered the mains connectors. Hard to see what was going on under the heat shrink. Hard to apply enough heat to get them off without lifting pads from the board. I'm going to have to do some repair here - find some new pins/connectors. It also looks like a previous battery has leaked and there is some corrosion in places. The part that had been used to replace the E505 in the past is marked 2789. Anyone recognize that? This was a disaster. Although fine on battery, and even with the battery disconnected, on AC, output voltage is millivolts(!) low. So I connect up my 3455A to the pre-supply output and the output is now correct. Poked around with the 'scope and same problem, as soon as I connect it, the problem goes away. I figure something is oscillating and connecting anything damps it. I can see the problem on the 10V output with the scope. 2mV jumps up and down. Not a square wave, but perhaps 25% duty cycle at 120Hz. I did eventually manage to catch a similar waveform at the emitter of the pass transistor. It would vary in amplitude - up to 40mV peak, dropping if I put my hand near the probe. After a couple of hours tinkering (try a zener that was 17.5V, capacitor across the zener etc.), I gave up and put the '2789' and original zener/2 1N4148s back in. The 10V output is back to normal, but the 1.018 setting is reading 1.0V. A careful inspection found that the grey wire had broken off the switch - a relatively easy fix. What next? Perhaps try the 18.1V zener - at least I can replace that from the top of the board. I don't feel like taking everything apart again to swap the '2789' out again. Orin. On Thu, Jun 20, 2013 at 1:49 PM, Bob Smither <smither@c-c-i.com> wrote: > On 06/17/2013 02:29 AM, Christopher Brown wrote: > > <snip> > > > That jfet and the shielded transformer are hard to get. Everything else > > on the power supply board is nothing special. All of the specially > > selected stuff is on the daughter board. > > <snip> > > I needed the current diode to repair a 731B as well. Try EBay.com. > Search for > "j505 current" and you will find them. > > -- > ========================================================================= > Bob Smither, PhD Circuit Concepts, Inc. > > "There is no worse tyranny than to force a man to pay for what > he does not want merely because you think it would be good for him." > -- Robert Heinlein > > Smither@C-C-I.Com http://www.C-C-I.Com 281-331-2744 > ========================================================================= > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
CB
Christopher Brown
Sun, Jun 23, 2013 9:27 PM

Do have have a shorting bar in place between guard and gnd?

One of the issues I tracked down on my worst unit was a leek from output

  • to guard.  Some of the tin had flaked off of the traces and formed a 6
    ohm connection from + PS out to guard.  Besides guard being hot I saw
    intermittent output.

Simply probing with a scope would return voltage to normal.

Simply placing a dry fingertip across the traces/guard in front of the
guard would do the same.

I asked about the shorting bar, because connecting guard to gnd in my
case would pull the ps output way down.

Anyway, worth checking for voltage on guard.

On 6/23/13 12:27 PM, Orin Eman wrote:

Well, an E505 and some 1n967Bs showed up yesterday.  I breadboarded them
with a 180 ohm resistor and a bench power supply that goes to 30V.

The 1N967Bs ranged from 16.8V to 18.1V when driven with the E505.  Voltage
across the resistor was about 0.195, so about 1.1mA.

I chose the 18.1V zener and started the fun and games of installing them.
Someone had soldered the mains connectors.  Hard to see what was going on
under the heat shrink.  Hard to apply enough heat to get them off without
lifting pads from the board.  I'm going to have to do some repair here -
find some new pins/connectors.  It also looks like a previous battery has
leaked and there is some corrosion in places.

The part that had been used to replace the E505 in the past is marked
2789.  Anyone recognize that?

This was a disaster.  Although fine on battery, and even with the battery
disconnected, on AC, output voltage is millivolts(!) low.  So I connect up
my 3455A to the pre-supply output and the output is now correct.  Poked
around with the 'scope and same problem, as soon as I connect it, the
problem goes away.  I figure something is oscillating and connecting
anything damps it.  I can see the problem on the 10V output with the
scope.  2mV jumps up and down.  Not a square wave, but perhaps 25% duty
cycle at 120Hz.  I did eventually manage to catch a similar waveform at the
emitter of the pass transistor.  It would vary in amplitude - up to 40mV
peak, dropping if I put my hand near the probe.

After a couple of hours tinkering (try a zener that was 17.5V, capacitor
across the zener etc.), I gave up and put the '2789' and original zener/2
1N4148s back in.  The 10V output is back to normal, but the 1.018 setting
is reading 1.0V.  A careful inspection found that the grey wire had broken
off the switch - a relatively easy fix.

What next?  Perhaps try the 18.1V zener - at least I can replace that from
the top of the board.  I don't feel like taking everything apart again to
swap the '2789' out again.

Orin.

On Thu, Jun 20, 2013 at 1:49 PM, Bob Smither smither@c-c-i.com wrote:

On 06/17/2013 02:29 AM, Christopher Brown wrote:

<snip>

That jfet and the shielded transformer are hard to get.  Everything else
on the power supply board is nothing special.  All of the specially
selected stuff is on the daughter board.

<snip>

I needed the current diode to repair a 731B as well.  Try EBay.com.
Search for
"j505 current" and you will find them.

--


---=======
Bob Smither, PhD                                  Circuit Concepts, Inc.

"There is no worse tyranny than to force a man to pay for what
he does not want merely because you think it would be good for him."
-- Robert Heinlein

Smither@C-C-I.Com            http://www.C-C-I.Com            281-331-2744


---=======


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Do have have a shorting bar in place between guard and gnd? One of the issues I tracked down on my worst unit was a leek from output + to guard. Some of the tin had flaked off of the traces and formed a 6 ohm connection from + PS out to guard. Besides guard being hot I saw intermittent output. Simply probing with a scope would return voltage to normal. Simply placing a _dry_ fingertip across the traces/guard in front of the guard would do the same. I asked about the shorting bar, because connecting guard to gnd in my case would pull the ps output _way down_. Anyway, worth checking for voltage on guard. On 6/23/13 12:27 PM, Orin Eman wrote: > Well, an E505 and some 1n967Bs showed up yesterday. I breadboarded them > with a 180 ohm resistor and a bench power supply that goes to 30V. > > The 1N967Bs ranged from 16.8V to 18.1V when driven with the E505. Voltage > across the resistor was about 0.195, so about 1.1mA. > > I chose the 18.1V zener and started the fun and games of installing them. > Someone had soldered the mains connectors. Hard to see what was going on > under the heat shrink. Hard to apply enough heat to get them off without > lifting pads from the board. I'm going to have to do some repair here - > find some new pins/connectors. It also looks like a previous battery has > leaked and there is some corrosion in places. > > The part that had been used to replace the E505 in the past is marked > 2789. Anyone recognize that? > > This was a disaster. Although fine on battery, and even with the battery > disconnected, on AC, output voltage is millivolts(!) low. So I connect up > my 3455A to the pre-supply output and the output is now correct. Poked > around with the 'scope and same problem, as soon as I connect it, the > problem goes away. I figure something is oscillating and connecting > anything damps it. I can see the problem on the 10V output with the > scope. 2mV jumps up and down. Not a square wave, but perhaps 25% duty > cycle at 120Hz. I did eventually manage to catch a similar waveform at the > emitter of the pass transistor. It would vary in amplitude - up to 40mV > peak, dropping if I put my hand near the probe. > > After a couple of hours tinkering (try a zener that was 17.5V, capacitor > across the zener etc.), I gave up and put the '2789' and original zener/2 > 1N4148s back in. The 10V output is back to normal, but the 1.018 setting > is reading 1.0V. A careful inspection found that the grey wire had broken > off the switch - a relatively easy fix. > > What next? Perhaps try the 18.1V zener - at least I can replace that from > the top of the board. I don't feel like taking everything apart again to > swap the '2789' out again. > > Orin. > > > > On Thu, Jun 20, 2013 at 1:49 PM, Bob Smither <smither@c-c-i.com> wrote: > >> On 06/17/2013 02:29 AM, Christopher Brown wrote: >> >> <snip> >> >>> That jfet and the shielded transformer are hard to get. Everything else >>> on the power supply board is nothing special. All of the specially >>> selected stuff is on the daughter board. >> >> <snip> >> >> I needed the current diode to repair a 731B as well. Try EBay.com. >> Search for >> "j505 current" and you will find them. >> >> -- >> ========================================================================= >> Bob Smither, PhD Circuit Concepts, Inc. >> >> "There is no worse tyranny than to force a man to pay for what >> he does not want merely because you think it would be good for him." >> -- Robert Heinlein >> >> Smither@C-C-I.Com http://www.C-C-I.Com 281-331-2744 >> ========================================================================= >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
OE
Orin Eman
Sun, Jun 23, 2013 11:27 PM

On Sun, Jun 23, 2013 at 2:27 PM, Christopher Brown cbrown@woods.net wrote:

Do have have a shorting bar in place between guard and gnd?

No, but it didn't make any difference if I shorted guard and either
terminal.

One of the issues I tracked down on my worst unit was a leek from output

  • to guard.  Some of the tin had flaked off of the traces and formed a 6
    ohm connection from + PS out to guard.  Besides guard being hot I saw
    intermittent output.

Simply probing with a scope would return voltage to normal.

Simply placing a dry fingertip across the traces/guard in front of the
guard would do the same.

I asked about the shorting bar, because connecting guard to gnd in my
case would pull the ps output way down.

Anyway, worth checking for voltage on guard.

Yes, there is some leakage to guard and a couple of V is readable, but the
leakage in the order of M ohms.  I did some checks with the DMM on ohms,
one lead on guard and ran the other probe around the power supply board and
there are definitely areas with measurable resistance.  I scrubbed some
with alcohol and electronics 'q-tips' and got the meter to read overload.
It looks like I'll have to take the bridge rectifier out to scrub under
it.  I was hoping I wouldn't have to do it, but I guess contamination from
a previous battery leak wasn't cleaned up well enough and I'll have to take
the board out.  Oh well, I can fix the broken pins at the same time.

The output voltage is behaving itself now though and holding steady at two
counts above 10V on the 3456A with 6-digit display selected.

Another thing I noticed is that the zener voltage has a lot of broadband
noise on it.  0.1uF or so across the zener makes it invisible on my 'scope.

Orin.

On Sun, Jun 23, 2013 at 2:27 PM, Christopher Brown <cbrown@woods.net> wrote: > Do have have a shorting bar in place between guard and gnd? > No, but it didn't make any difference if I shorted guard and either terminal. > > One of the issues I tracked down on my worst unit was a leek from output > + to guard. Some of the tin had flaked off of the traces and formed a 6 > ohm connection from + PS out to guard. Besides guard being hot I saw > intermittent output. > > Simply probing with a scope would return voltage to normal. > > Simply placing a _dry_ fingertip across the traces/guard in front of the > guard would do the same. > > I asked about the shorting bar, because connecting guard to gnd in my > case would pull the ps output _way down_. > > Anyway, worth checking for voltage on guard. > > Yes, there is some leakage to guard and a couple of V is readable, but the leakage in the order of M ohms. I did some checks with the DMM on ohms, one lead on guard and ran the other probe around the power supply board and there are definitely areas with measurable resistance. I scrubbed some with alcohol and electronics 'q-tips' and got the meter to read overload. It looks like I'll have to take the bridge rectifier out to scrub under it. I was hoping I wouldn't have to do it, but I guess contamination from a previous battery leak wasn't cleaned up well enough and I'll have to take the board out. Oh well, I can fix the broken pins at the same time. The output voltage is behaving itself now though and holding steady at two counts above 10V on the 3456A with 6-digit display selected. Another thing I noticed is that the zener voltage has a lot of broadband noise on it. 0.1uF or so across the zener makes it invisible on my 'scope. Orin.