Speaking of the 03458-66529 reference board added to the HP 3458A by Fluke (before Fluke offered the 8508A), has anyone ever seen a schematic of that board or how Fluke could rate that reference at 2 ppm? One person on this forum theorized that by lowering the operating temperature of the LTZ1000A to around 70 degrees C, which is what the LTC device specification note recommends, this might have what was done to achieve the better drift specification. It was noted that HP was running the LTZ1000A at around 105 degrees C which might accelerate the aging rate. I lowered my references to around 70 C and they have seemed to be a little more stable when checked against the Fluke 732A DC Reference Standard. But I never started a project to follow up on this theory. Just another project to play with when I "get a round toit"
Bill
I think the main thing they do is age and screen.
On Tue, Oct 30, 2012 at 6:48 AM, Bill Gold wpgold3637@att.net wrote:
Speaking of the 03458-66529 reference board added to the HP 3458A by Fluke
(before Fluke offered the 8508A), has anyone ever seen a schematic of that
board or how Fluke could rate that reference at 2 ppm? One person on this
forum theorized that by lowering the operating temperature of the LTZ1000A
to around 70 degrees C, which is what the LTC device specification note
recommends, this might have what was done to achieve the better drift
specification. It was noted that HP was running the LTZ1000A at around 105
degrees C which might accelerate the aging rate. I lowered my references
to around 70 C and they have seemed to be a little more stable when checked
against the Fluke 732A DC Reference Standard. But I never started a
project to follow up on this theory. Just another project to play with
when I "get a round toit"
Bill
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
--
John Phillips
Per my conversation with Gary Biermann, the 4 ppm and the 2 ppm units are
the same as the 8 ppm units except they perform better. Purely a selection
of best performers and nothing to do with a special circuit.
I don't know about Fluke.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of John Phillips
Sent: Tuesday, October 30, 2012 11:07 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] (New to me) 3458A Trivia
I think the main thing they do is age and screen.
On Tue, Oct 30, 2012 at 6:48 AM, Bill Gold wpgold3637@att.net wrote:
Speaking of the 03458-66529 reference board added to the HP 3458A by
Fluke (before Fluke offered the 8508A), has anyone ever seen a
schematic of that board or how Fluke could rate that reference at 2
ppm? One person on this forum theorized that by lowering the
operating temperature of the LTZ1000A to around 70 degrees C, which is
what the LTC device specification note recommends, this might have
what was done to achieve the better drift specification. It was noted
that HP was running the LTZ1000A at around 105 degrees C which might
accelerate the aging rate. I lowered my references to around 70 C and
they have seemed to be a little more stable when checked against the
Fluke 732A DC Reference Standard. But I never started a project to
follow up on this theory. Just another project to play with when I "get a
round toit"
Bill