Well I finally got the 615nm Led's installed and the results seem so so.
The zero drifted was significantly on the 1uV range. So I used two
LT1634 1.25 Volt references with the current set by two 43k resistors
from the ±15 volt supplies. The previous reference used came from the
-0.4 volts at TP1 and +0.6 volts at TP8. Removed the 150k resistor at
the positive end of the zero control and lifted the negative end from
the circuit board the connected the ±1.25 ref to the zero pot. Much
better stability, drifts by noise about ±.1 to ±.2uv now on 1uv range.
Is this normal now? Never used a new Fluke 845ab before. The white led's
seemed to filter the noise better with its persistent, may go back. The
zero seems stable with ±2 F change in room temperature now.
Dallas:
I check my 845ab yesterday, S/N 4470002. It still has the neon
bulbs/photo resistors chopper.
From what I can observe the Peak to Peak noise on the 1 mv range is
about +/- 0.1 mv, or a total of around 0.2 mv. As far as temperature drift
I don't know. Never tried to measure that parameter.
Sure would love to know if the FLUKE modification to the 845ab made in
1993, when the new circuit was put into production, makes the noise lower.
Or better, if your efforts can make the noise lower on the 1 mv range.
Bill
----- Original Message -----
From: "Dallas Smith" dosmith@outlook.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Wednesday, September 24, 2014 12:06 PM
Subject: [volt-nuts] 845ab Led Mod Update
Well I finally got the 615nm Led's installed and the results seem so so.
The zero drifted was significantly on the 1uV range. So I used two
LT1634 1.25 Volt references with the current set by two 43k resistors
from the ±15 volt supplies. The previous reference used came from the
-0.4 volts at TP1 and +0.6 volts at TP8. Removed the 150k resistor at
the positive end of the zero control and lifted the negative end from
the circuit board the connected the ±1.25 ref to the zero pot. Much
better stability, drifts by noise about ±.1 to ±.2uv now on 1uv range.
Is this normal now? Never used a new Fluke 845ab before. The white led's
seemed to filter the noise better with its persistent, may go back. The
zero seems stable with ±2 F change in room temperature now.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
and follow the instructions there.
Sorry I should have said "uv" where ever I said "mv". Too early in the
morning where I live in the world.
Bill
----- Original Message -----
From: "Bill Gold" wpgold3637@att.net
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Thursday, September 25, 2014 8:10 AM
Subject: Re: [volt-nuts] 845ab Led Mod Update
Dallas:
I check my 845ab yesterday, S/N 4470002. It still has the neon
bulbs/photo resistors chopper.
From what I can observe the Peak to Peak noise on the 1 mv range is
about +/- 0.1 mv, or a total of around 0.2 mv. As far as temperature
drift
I don't know. Never tried to measure that parameter.
Sure would love to know if the FLUKE modification to the 845ab made in
1993, when the new circuit was put into production, makes the noise lower.
Or better, if your efforts can make the noise lower on the 1 mv range.
Bill
----- Original Message -----
From: "Dallas Smith" dosmith@outlook.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Wednesday, September 24, 2014 12:06 PM
Subject: [volt-nuts] 845ab Led Mod Update
Well I finally got the 615nm Led's installed and the results seem so so.
The zero drifted was significantly on the 1uV range. So I used two
LT1634 1.25 Volt references with the current set by two 43k resistors
from the ±15 volt supplies. The previous reference used came from the
-0.4 volts at TP1 and +0.6 volts at TP8. Removed the 150k resistor at
the positive end of the zero control and lifted the negative end from
the circuit board the connected the ±1.25 ref to the zero pot. Much
better stability, drifts by noise about ±.1 to ±.2uv now on 1uv range.
Is this normal now? Never used a new Fluke 845ab before. The white led's
seemed to filter the noise better with its persistent, may go back. The
zero seems stable with ±2 F change in room temperature now.