Charlie,
Thanks for the data. Before I can make any intelligent comments about the
stability of my 'system', I am going to have to solve the HPIB issue and
find a way to communicate with the 3458A. I solved RS232 with the 7081.
Too bad the 3458A doesn't have RS232.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Charles Black
Sent: Saturday, October 29, 2011 2:13 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 3458A Questions - The Sequel
Hi Joe,
Your noise and drift on your 731B is more than I see on my Datron 4910.
I have no experience with the 731B but my 4910 is very similar in
specifications to a 732B. My major drift with my 3458a measuring a 4910
cell is the room temperature. It is about 10 microvolt per degree C. My
peak to peak noise is dependent on NPLC settings up to 200 where it
becomes very usable with only a minor improvement at NPLC = 1000. NPLC
1000 is useful to get the ultimate accuracy on my system since it
integrates out some of the lowest frequency noise in the 4910. I use
shielded #24 cable with no connectors. The wire is tin plated Tefzel
STP but that seems to only cause 0 to 10 nv errors and reaches thermal
stability much faster than my otherwise identical #22 STP. I have
included two recent plots of one of my Datron 4910 Cells (it has four
cells which perform similarly) to illustrate NPLC settings. Notice the
downward drift in both plots that is caused by changes in room
temperature and the peak ti peak noise which I assume is 4910 noise. The
4910 noise spec is 0.04 PPM RMS and my cells are well within that.
I tried using connectors but only the gold plated spade lugs were af any
use to me. Most of my issues had to to do with thermal mass. To use
banana plugs I had to wait and wait. Dual gold plated banana plugs
(about 1mv thermal offset at first) never stabilize on my 3458A. I have
too many air currents I guess. Fortunately no connectors works perfectly
here and is the cheapest too.
Enjoy your new (to you) 3458A.
Charlie
Hi Joe,
The RS-232 bus is easy cheap and reliable but too slow for the 3458a.
The 3458A can send a lot of data over the bus very quickly so at the
time the IEEE bus was an easy choice.
I bought the IEEE-488 PCI card on eeBay and also a long long cable.
Later I decided to spring for a GPIB-ENET/100 for $200 to free up the
single PCI slot in my PC and to be able to be further away with out the
long IEEE-488 cable. Both were National products and worked well once
the manual was thoroughly read.
I use National Instruments National Instruments Measurement and
Instrumentation Explorer which is a free download from NI which is part
of the NI VISA install. It can see your IEEE-488 devices and let you
send text commands to them. It also has the Spy program that lets you
monitor the actual text sent and received in real time. Just remember to
alter the ID command test from the *IDN? to ID?. Also "BEEP" works. Case
doesn't matter. You will get a communication error if you don't set
"END" to ALWAYS in the 3458a menu.
On 10/29/2011 12:23 PM, J. L. Trantham wrote:
Charlie,
Thanks for the data. Before I can make any intelligent comments about the
stability of my 'system', I am going to have to solve the HPIB issue and
find a way to communicate with the 3458A. I solved RS232 with the 7081.
Too bad the 3458A doesn't have RS232.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Charles Black
Sent: Saturday, October 29, 2011 2:13 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 3458A Questions - The Sequel
Hi Joe,
Your noise and drift on your 731B is more than I see on my Datron 4910.
I have no experience with the 731B but my 4910 is very similar in
specifications to a 732B. My major drift with my 3458a measuring a 4910
cell is the room temperature. It is about 10 microvolt per degree C. My
peak to peak noise is dependent on NPLC settings up to 200 where it
becomes very usable with only a minor improvement at NPLC = 1000. NPLC
1000 is useful to get the ultimate accuracy on my system since it
integrates out some of the lowest frequency noise in the 4910. I use
shielded #24 cable with no connectors. The wire is tin plated Tefzel
STP but that seems to only cause 0 to 10 nv errors and reaches thermal
stability much faster than my otherwise identical #22 STP. I have
included two recent plots of one of my Datron 4910 Cells (it has four
cells which perform similarly) to illustrate NPLC settings. Notice the
downward drift in both plots that is caused by changes in room
temperature and the peak ti peak noise which I assume is 4910 noise. The
4910 noise spec is 0.04 PPM RMS and my cells are well within that.
I tried using connectors but only the gold plated spade lugs were af any
use to me. Most of my issues had to to do with thermal mass. To use
banana plugs I had to wait and wait. Dual gold plated banana plugs
(about 1mv thermal offset at first) never stabilize on my 3458A. I have
too many air currents I guess. Fortunately no connectors works perfectly
here and is the cheapest too.
Enjoy your new (to you) 3458A.
Charlie
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Charles Black
Sent: Saturday, October 29, 2011 3:21 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 3458A Questions - The Sequel
Hi Joe,
The RS-232 bus is easy cheap and reliable but too slow for the 3458a.
The 3458A can send a lot of data over the bus very quickly so at the
time the IEEE bus was an easy choice.
I bought the IEEE-488 PCI card on eeBay and also a long long cable.
Later I decided to spring for a GPIB-ENET/100 for $200 to free up the
single PCI slot in my PC and to be able to be further away with out the
long IEEE-488 cable. Both were National products and worked well once
the manual was thoroughly read.
I use National Instruments National Instruments Measurement and
Instrumentation Explorer which is a free download from NI which is part
of the NI VISA install. It can see your IEEE-488 devices and let you
send text commands to them. It also has the Spy program that lets you
monitor the actual text sent and received in real time. Just remember to
alter the ID command test from the *IDN? to ID?. Also "BEEP" works. Case
doesn't matter. You will get a communication error if you don't set
"END" to ALWAYS in the 3458a menu.
Charlie,
I got out my NI GPIB-USB-HS connector, loaded my NI 488.2 on my XP laptop,
and now have been able to communicate with the meter. Thanks very much for
the tip of setting 'END' to 'ALWAYS'. Until I did that, nothing!
Next, I would like to read the contents of the EPROM and read the
calibration settings. So far, I have not been able to solve that.
When I send or 'query' or 'read' 'CALRAM?', all I get back is the current
reading on the meter as it measures my Fluke 731B.
More work to do.
Joe
In message 8E6680B14BDB467798C413BB55F193C7@cardiac5f772ce, "J. L. Trantham"
writes:
Next, I would like to read the contents of the EPROM and read the
calibration settings. So far, I have not been able to solve that.
You need to do direct memory reads for that.
Use the
MREAD? <adr>
command for that, it reads 16 bits at a time.
The EPROM is at address 0...393214 [0x00000...0x5fffe]
The NVRAM is at address 393216...397310 [0x60000...0x60ffe] but only
the high byte in each word is valid.
--