Joe wrote:
Just messing around, I measured a used 9 V battery. On the 100 uV
scale and with some interprolating of the last dial, I read 9.189795
V at null (with a tiny bit of wobble on the needle).
Reading an 895A to 7 significant figures is a case of "false
precision," which can lead to overconfidence in the accuracy of
measurement results. Fluke specifies the accuracy of the 895A in
differential mode as +/- (0.0025% of input + 0.0001% of range + 5 uV)
at calibration temperature (around +/- 250 uV uncertainty for the
measurement you describe), with calibration recommended every 30 days
to maintain that accuracy.
Best regards,
Charles
You can get two digits from the last dial scale. I didn't say I believed
them :-) In fact, given the age and history of this thing, I'm surprised
that even the first digit is believeable.
Joe Gray
W5JG
On Wed, Aug 14, 2013 at 12:43 AM, Charles Steinmetz
csteinmetz@yandex.comwrote:
Joe wrote:
Just messing around, I measured a used 9 V battery. On the 100 uV scale
and with some interprolating of the last dial, I read 9.189795 V at null
(with a tiny bit of wobble on the needle).
Reading an 895A to 7 significant figures is a case of "false precision,"
which can lead to overconfidence in the accuracy of measurement results.
Fluke specifies the accuracy of the 895A in differential mode as +/-
(0.0025% of input + 0.0001% of range + 5 uV) at calibration temperature
(around +/- 250 uV uncertainty for the measurement you describe), with
calibration recommended every 30 days to maintain that accuracy.
Best regards,
Charles