Joe wrote:

Just messing around, I measured a used 9 V battery. On the 100 uV
scale and with some interprolating of the last dial, I read 9.189795
V at null (with a tiny bit of wobble on the needle).

Reading an 895A to 7 significant figures is a case of "false
precision," which can lead to overconfidence in the accuracy of
measurement results.  Fluke specifies the accuracy of the 895A in
differential mode as +/- (0.0025% of input + 0.0001% of range + 5 uV)
at calibration temperature (around +/- 250 uV uncertainty for the
measurement you describe), with calibration recommended every 30 days
to maintain that accuracy.

Best regards,

Charles

Joe wrote: >Just messing around, I measured a used 9 V battery. On the 100 uV >scale and with some interprolating of the last dial, I read 9.189795 >V at null (with a tiny bit of wobble on the needle). Reading an 895A to 7 significant figures is a case of "false precision," which can lead to overconfidence in the accuracy of measurement results. Fluke specifies the accuracy of the 895A in differential mode as +/- (0.0025% of input + 0.0001% of range + 5 uV) at calibration temperature (around +/- 250 uV uncertainty for the measurement you describe), with calibration recommended every 30 days to maintain that accuracy. Best regards, Charles
JG
Joseph Gray
Wed, Aug 14, 2013 1:13 PM

You can get two digits from the last dial scale. I didn't say I believed
them :-) In fact, given the age and history of this thing, I'm surprised
that even the first digit is believeable.

Joe Gray
W5JG

On Wed, Aug 14, 2013 at 12:43 AM, Charles Steinmetz
csteinmetz@yandex.comwrote:

Joe wrote:

Just messing around, I measured a used 9 V battery. On the 100 uV scale

and with some interprolating of the last dial, I read 9.189795 V at null
(with a tiny bit of wobble on the needle).

Reading an 895A to 7 significant figures is a case of "false precision,"
which can lead to overconfidence in the accuracy of measurement results.
Fluke specifies the accuracy of the 895A in differential mode as +/-
(0.0025% of input + 0.0001% of range + 5 uV) at calibration temperature
(around +/- 250 uV uncertainty for the measurement you describe), with
calibration recommended every 30 days to maintain that accuracy.

Best regards,

Charles

_____________**
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/**
mailman/listinfo/volt-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

You can get two digits from the last dial scale. I didn't say I believed them :-) In fact, given the age and history of this thing, I'm surprised that even the first digit is believeable. Joe Gray W5JG On Wed, Aug 14, 2013 at 12:43 AM, Charles Steinmetz <csteinmetz@yandex.com>wrote: > Joe wrote: > > Just messing around, I measured a used 9 V battery. On the 100 uV scale >> and with some interprolating of the last dial, I read 9.189795 V at null >> (with a tiny bit of wobble on the needle). >> > > Reading an 895A to 7 significant figures is a case of "false precision," > which can lead to overconfidence in the accuracy of measurement results. > Fluke specifies the accuracy of the 895A in differential mode as +/- > (0.0025% of input + 0.0001% of range + 5 uV) at calibration temperature > (around +/- 250 uV uncertainty for the measurement you describe), with > calibration recommended every 30 days to maintain that accuracy. > > Best regards, > > Charles > > > > ______________________________**_________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/** > mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts> > and follow the instructions there. > >