Interpreting the ADEV plot in TimeLab

PA
Paul A. Cianciolo
Fri, Oct 21, 2011 10:36 PM

Hello,

I use  TimeLab and the HP5370B and a Prologix GPIB interface.
After making several 10 Hr runs with the same oscillator I noticed something

When laying the successive runs on top of each other there are little blips
and bumps that line up perfectly on top of each other.
For example each run has a small blip at 10 seconds in, and a rise from 50
seconds to 150 seconds

This was happening when I was using the NI PCM-CIA interface as well.

If time is the X axis and variation of frequency is the y axis , how can
something repeatedly  happen at  2 points in time.
I am sampling at  1 S per sec and 10 Hours duration in these particular
traces.

Can I have some sort of cabling issue?  Or is there something in ADEV
algorithm that forces this to happen.?

Thank you

Paul A. Cianciolo
W1VLF
http://www.rescueelectronics.com/
Our business computer network is  powered exclusively by solar and wind
power.
Converting Photons to Electrons for over 20 years

Hello, I use TimeLab and the HP5370B and a Prologix GPIB interface. After making several 10 Hr runs with the same oscillator I noticed something When laying the successive runs on top of each other there are little blips and bumps that line up perfectly on top of each other. For example each run has a small blip at 10 seconds in, and a rise from 50 seconds to 150 seconds This was happening when I was using the NI PCM-CIA interface as well. If time is the X axis and variation of frequency is the y axis , how can something repeatedly happen at 2 points in time. I am sampling at 1 S per sec and 10 Hours duration in these particular traces. Can I have some sort of cabling issue? Or is there something in ADEV algorithm that forces this to happen.? Thank you Paul A. Cianciolo W1VLF http://www.rescueelectronics.com/ Our business computer network is powered exclusively by solar and wind power. Converting Photons to Electrons for over 20 years
JM
John Miles
Sat, Oct 22, 2011 1:58 AM

Hello,

I use  TimeLab and the HP5370B and a Prologix GPIB interface.
After making several 10 Hr runs with the same oscillator I noticed

something

When laying the successive runs on top of each other there are little

blips

and bumps that line up perfectly on top of each other.
For example each run has a small blip at 10 seconds in, and a rise from 50
seconds to 150 seconds

This was happening when I was using the NI PCM-CIA interface as well.

If time is the X axis and variation of frequency is the y axis , how can
something repeatedly  happen at  2 points in time.

The x-axis should be thought of as an indication of period, not absolute
time.  The whole idea behind the ADEV plot is to tell you how accurate the
clock can be expected to be when observed at recurring intervals of T
seconds, where T is any given tau value on the X-axis.

This can be confusing because when you watch the ADEV trace appear over
time, it seems to advance at a rate reminiscent of the time label on the
X-axis.  In reality this is because the graph can only be updated for a
given tau interval once enough samples are available at that timescale.

Say you' re checking your watch for accuracy on a day-to-day basis
(tau=86400 seconds).  You couldn't make an accurate judgment if you only
checked it after a single day.  You wouldn't feel confident in your t=86400s
accuracy estimate until you had checked it every day for a week, for
instance.  Even then, you might observe that your watch gains an average of
2 seconds per day the first week, 3 the next week, and 8 the week after
that.  For this reason, you'd need to run your ADEV test for several days if
you want to see what the graph looks like at 86400 seconds... and the error
is likely to be large when that part of the graph first appears.

In your case, if you look at the phase and/or frequency difference plots you
may be able to discern some kind of regular variation or structure that
recurs every 10 seconds.  An ADEV test that runs long enough will often
exhibit small bumps corresponding to HVAC cycle times, day/night temperature
changes, and other effects.  If you were sampling fast enough, you'd also
tend to see bumps at the power-line frequency and its first few harmonics.

-- john

> Hello, > > I use TimeLab and the HP5370B and a Prologix GPIB interface. > After making several 10 Hr runs with the same oscillator I noticed something > > When laying the successive runs on top of each other there are little blips > and bumps that line up perfectly on top of each other. > For example each run has a small blip at 10 seconds in, and a rise from 50 > seconds to 150 seconds > > This was happening when I was using the NI PCM-CIA interface as well. > > If time is the X axis and variation of frequency is the y axis , how can > something repeatedly happen at 2 points in time. The x-axis should be thought of as an indication of period, not absolute time. The whole idea behind the ADEV plot is to tell you how accurate the clock can be expected to be when observed at recurring intervals of T seconds, where T is any given tau value on the X-axis. This can be confusing because when you watch the ADEV trace appear over time, it seems to advance at a rate reminiscent of the time label on the X-axis. In reality this is because the graph can only be updated for a given tau interval once enough samples are available at that timescale. Say you' re checking your watch for accuracy on a day-to-day basis (tau=86400 seconds). You couldn't make an accurate judgment if you only checked it after a single day. You wouldn't feel confident in your t=86400s accuracy estimate until you had checked it every day for a week, for instance. Even then, you might observe that your watch gains an average of 2 seconds per day the first week, 3 the next week, and 8 the week after that. For this reason, you'd need to run your ADEV test for several days if you want to see what the graph looks like at 86400 seconds... and the error is likely to be large when that part of the graph first appears. In your case, if you look at the phase and/or frequency difference plots you may be able to discern some kind of regular variation or structure that recurs every 10 seconds. An ADEV test that runs long enough will often exhibit small bumps corresponding to HVAC cycle times, day/night temperature changes, and other effects. If you were sampling fast enough, you'd also tend to see bumps at the power-line frequency and its first few harmonics. -- john