Just after the groups collective opinion on the suitability of a fluke 332d
as a voltage standard for a home lab.
This email is free from viruses and malware because avast! Antivirus protection is active.
http://www.avast.com
Ken,
I think the 332D would be a good tool for calibration of DMM's, etc., but
probably a bit of an energy hog to leave turned on all the time.
What do you want to do?
For a constantly on 10 VDC reference, I would consider a 731B or 732A. For
a 'calibrator', the 332D would be a good DC voltage calibrator up to 1100
VDC, IIRC. If you also need AC voltage, current, and resistance, consider a
Fluke 5100B. For separate units, consider a 5200 for AC voltage and 5450 or
some individual standard resistors for resistance. If you want only DC
voltage and more stability than the 332D, consider a 5440B.
Careful, this stuff can be addicting.
Good luck.
Joe