What exactly would be the extended characteristics on "old" cell that
render it no longer useful?
Dave
NR1Dx
On 9/22/2011 8:32 AM, Chuck Harris wrote:
As I understand it, an unsaturated cell that measures 1.0178V is many,
many years
past being useful as a standard, as is any unsaturated cell that
measures below
1.0183V.
Saturated cells are very rare outside of laboratory conditions, as
inverting the cell
will ruin it. Unsaturated cells can withstand the topsy turvy travel
conditions
presented by shipping.
-Chuck Harris
J. L. Trantham wrote:
Roy,
The 'experts' are here on the list. From my reading, they have about
a 50 uV/Yr
expected decline in value, thus the need for periodic 'calibration'.
I do not have a
feel for useful life expectancy but the two I have are in good
condition and only the
latest one has a value on the front of '1.01931 @ 20 degrees C' with
a date on the
side of '2-4-69'. This AM, it measures 1.017894. However, the 7081 is
not known to be
accurate, yet.
The other cell still measures 1.018709 this AM.
So, it would appear that they will make a good standard but will need
to be measured
from time to time to generate a log of date, temperature, and value.
I wonder how long it has to be 'still' before it is again stable?
Joe
----- Original Message ----- From: "Roy Phillips"
phill.r1@btinternet.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Thursday, September 22, 2011 5:16 AM
Subject: Re: [volt-nuts] Eppley Standard Cell and What Is It?
Joe
Interesting point regarding the "Standard Cell", I have one that was
purchased some
12 years ago made by Western. It must be well beyond its calibrated
accuracy. It
currently reads 1.018679 volts according to my 3458A. It seems to be
stable but
below its original labeled standard (label lost). Are they worth
keeping ?
Roy
From: "J. L. Trantham" jltran@att.net
Sent: Thursday, September 22, 2011 2:37 AM
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Subject: [volt-nuts] Eppley Standard Cell and What Is It?
I purchased an eBay item (190538109632) for the 'Standard Cell'
that included
another item.
What is it?
The 'Standard Cell' is an 'Eppley Laboratory Catalog Number 100 (SN
762058) Low
Termperature Coefficient Standard of E. M. F.' that has written on
it '1.01931 @
20 degree C' with a date of what I think is '2-4-69' which I think
means Feburary
4, 1969.
Using my 'new' Solartron 7081, I read the voltage at 1.017844 +/-
about 1 uV. If I
have the math correct, that is a decrease of about 35 uV/yr which,
I think, is
about what is predicted for these cell's, assuming my 7081 is
accurate.
I also have another Eppley Standard Cell (SN 821855) that reads
1.018709 but no
date or prior reading on the unit.
On another, unrelated note, I have been able to chase down the US
branch of Ametek
in Oak Ridge, TN, and they will be getting my 7081 back to send to
England for
calibration and 'service best effort', given that the meter ceased
production in
1998 with end of support 2008. TAT is about 4 weeks in England.
Now, what is that other thing in the listing?
Joe
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Chuck
What exactly would be the extended characteristics on "old" cell that
render it no longer useful?
Dave
NR1Dx
On 9/22/2011 8:32 AM, Chuck Harris wrote:
> As I understand it, an unsaturated cell that measures 1.0178V is many,
> many years
> past being useful as a standard, as is any unsaturated cell that
> measures below
> 1.0183V.
>
> Saturated cells are very rare outside of laboratory conditions, as
> inverting the cell
> will ruin it. Unsaturated cells can withstand the topsy turvy travel
> conditions
> presented by shipping.
>
> -Chuck Harris
>
> J. L. Trantham wrote:
>> Roy,
>>
>> The 'experts' are here on the list. From my reading, they have about
>> a 50 uV/Yr
>> expected decline in value, thus the need for periodic 'calibration'.
>> I do not have a
>> feel for useful life expectancy but the two I have are in good
>> condition and only the
>> latest one has a value on the front of '1.01931 @ 20 degrees C' with
>> a date on the
>> side of '2-4-69'. This AM, it measures 1.017894. However, the 7081 is
>> not known to be
>> accurate, yet.
>>
>> The other cell still measures 1.018709 this AM.
>>
>> So, it would appear that they will make a good standard but will need
>> to be measured
>> from time to time to generate a log of date, temperature, and value.
>>
>> I wonder how long it has to be 'still' before it is again stable?
>>
>> Joe
>>
>>
>> ----- Original Message ----- From: "Roy Phillips"
>> <phill.r1@btinternet.com>
>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>> Sent: Thursday, September 22, 2011 5:16 AM
>> Subject: Re: [volt-nuts] Eppley Standard Cell and What Is It?
>>
>>
>>> Joe
>>> Interesting point regarding the "Standard Cell", I have one that was
>>> purchased some
>>> 12 years ago made by Western. It must be well beyond its calibrated
>>> accuracy. It
>>> currently reads 1.018679 volts according to my 3458A. It seems to be
>>> stable but
>>> below its original labeled standard (label lost). Are they worth
>>> keeping ?
>>> Roy
>>>
>>>
>>> --------------------------------------------------
>>> From: "J. L. Trantham" <jltran@att.net>
>>> Sent: Thursday, September 22, 2011 2:37 AM
>>> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
>>> Subject: [volt-nuts] Eppley Standard Cell and What Is It?
>>>
>>>> I purchased an eBay item (190538109632) for the 'Standard Cell'
>>>> that included
>>>> another item.
>>>>
>>>> What is it?
>>>>
>>>> The 'Standard Cell' is an 'Eppley Laboratory Catalog Number 100 (SN
>>>> 762058) Low
>>>> Termperature Coefficient Standard of E. M. F.' that has written on
>>>> it '1.01931 @
>>>> 20 degree C' with a date of what I think is '2-4-69' which I think
>>>> means Feburary
>>>> 4, 1969.
>>>>
>>>> Using my 'new' Solartron 7081, I read the voltage at 1.017844 +/-
>>>> about 1 uV. If I
>>>> have the math correct, that is a decrease of about 35 uV/yr which,
>>>> I think, is
>>>> about what is predicted for these cell's, assuming my 7081 is
>>>> accurate.
>>>>
>>>> I also have another Eppley Standard Cell (SN 821855) that reads
>>>> 1.018709 but no
>>>> date or prior reading on the unit.
>>>>
>>>> On another, unrelated note, I have been able to chase down the US
>>>> branch of Ametek
>>>> in Oak Ridge, TN, and they will be getting my 7081 back to send to
>>>> England for
>>>> calibration and 'service best effort', given that the meter ceased
>>>> production in
>>>> 1998 with end of support 2008. TAT is about 4 weeks in England.
>>>>
>>>> Now, what is that other thing in the listing?
>>>>
>>>> Joe
>>>> _______________________________________________
>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>> and follow the instructions there.
>>>>
>>>
>>>
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>>
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
--
Dave& Lynn Henderson
Manuals@ArtekMedia.com
www.Artekmedia.com
PO Box 175
Welch,MN 55089
PL
Pete Lancashire
Thu, Sep 22, 2011 3:24 PM
phil sez: Are they worth keeping ?
even if not practical, time to get out the cleaners, polish, etc. and
put it on the mantel
or other display of treasures. any {volt,time,location,etc}-nut must
have such a spot
or two.
-pete
On Thu, Sep 22, 2011 at 3:16 AM, Roy Phillips phill.r1@btinternet.com wrote:
Joe
Interesting point regarding the "Standard Cell", I have one that was
purchased some 12 years ago made by Western. It must be well beyond its
calibrated accuracy. It currently reads 1.018679 volts according to my
3458A. It seems to be stable but below its original labeled standard (label
lost). Are they worth keeping ?
Roy
From: "J. L. Trantham" jltran@att.net
Sent: Thursday, September 22, 2011 2:37 AM
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Subject: [volt-nuts] Eppley Standard Cell and What Is It?
I purchased an eBay item (190538109632) for the 'Standard Cell' that
included another item.
What is it?
The 'Standard Cell' is an 'Eppley Laboratory Catalog Number 100 (SN
762058) Low Termperature Coefficient Standard of E. M. F.' that has written
on it '1.01931 @ 20 degree C' with a date of what I think is '2-4-69' which
I think means Feburary 4, 1969.
Using my 'new' Solartron 7081, I read the voltage at 1.017844 +/- about 1
uV. If I have the math correct, that is a decrease of about 35 uV/yr which,
I think, is about what is predicted for these cell's, assuming my 7081 is
accurate.
I also have another Eppley Standard Cell (SN 821855) that reads 1.018709
but no date or prior reading on the unit.
On another, unrelated note, I have been able to chase down the US branch
of Ametek in Oak Ridge, TN, and they will be getting my 7081 back to send to
England for calibration and 'service best effort', given that the meter
ceased production in 1998 with end of support 2008. TAT is about 4 weeks in
England.
Now, what is that other thing in the listing?
Joe
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
phil sez: Are they worth keeping ?
even if not practical, time to get out the cleaners, polish, etc. and
put it on the mantel
or other display of treasures. any {volt,time,location,etc}-nut must
have such a spot
or two.
-pete
On Thu, Sep 22, 2011 at 3:16 AM, Roy Phillips <phill.r1@btinternet.com> wrote:
> Joe
> Interesting point regarding the "Standard Cell", I have one that was
> purchased some 12 years ago made by Western. It must be well beyond its
> calibrated accuracy. It currently reads 1.018679 volts according to my
> 3458A. It seems to be stable but below its original labeled standard (label
> lost). Are they worth keeping ?
> Roy
>
>
> --------------------------------------------------
> From: "J. L. Trantham" <jltran@att.net>
> Sent: Thursday, September 22, 2011 2:37 AM
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Subject: [volt-nuts] Eppley Standard Cell and What Is It?
>
>> I purchased an eBay item (190538109632) for the 'Standard Cell' that
>> included another item.
>>
>> What is it?
>>
>> The 'Standard Cell' is an 'Eppley Laboratory Catalog Number 100 (SN
>> 762058) Low Termperature Coefficient Standard of E. M. F.' that has written
>> on it '1.01931 @ 20 degree C' with a date of what I think is '2-4-69' which
>> I think means Feburary 4, 1969.
>>
>> Using my 'new' Solartron 7081, I read the voltage at 1.017844 +/- about 1
>> uV. If I have the math correct, that is a decrease of about 35 uV/yr which,
>> I think, is about what is predicted for these cell's, assuming my 7081 is
>> accurate.
>>
>> I also have another Eppley Standard Cell (SN 821855) that reads 1.018709
>> but no date or prior reading on the unit.
>>
>> On another, unrelated note, I have been able to chase down the US branch
>> of Ametek in Oak Ridge, TN, and they will be getting my 7081 back to send to
>> England for calibration and 'service best effort', given that the meter
>> ceased production in 1998 with end of support 2008. TAT is about 4 weeks in
>> England.
>>
>> Now, what is that other thing in the listing?
>>
>> Joe
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
CH
Chuck Harris
Thu, Sep 22, 2011 3:30 PM
I don't know exactly, but I suspect that the active materials are
expended, or are sequestered, in such a way that they are not available.
That would cause the impedance to rise dramatically, and make the reading
unstable.
-Chuck Harris
Artekmedia wrote:
Chuck
What exactly would be the extended characteristics on "old" cell that render it no
longer useful?
Dave
NR1Dx
I don't know exactly, but I suspect that the active materials are
expended, or are sequestered, in such a way that they are not available.
That would cause the impedance to rise dramatically, and make the reading
unstable.
-Chuck Harris
Artekmedia wrote:
> Chuck
>
> What exactly would be the extended characteristics on "old" cell that render it no
> longer useful?
>
> Dave
> NR1Dx
CH
Chuck Harris
Thu, Sep 22, 2011 3:31 PM
The leads that go through the glass are platinum, that alone makes
them worth keeping.
Pete Lancashire wrote:
phil sez: Are they worth keeping ?
even if not practical, time to get out the cleaners, polish, etc. and
put it on the mantel
or other display of treasures. any {volt,time,location,etc}-nut must
have such a spot
or two.
-pete
The leads that go through the glass are platinum, that alone makes
them worth keeping.
Pete Lancashire wrote:
> phil sez: Are they worth keeping ?
>
> even if not practical, time to get out the cleaners, polish, etc. and
> put it on the mantel
> or other display of treasures. any {volt,time,location,etc}-nut must
> have such a spot
> or two.
>
> -pete