I've watched a few of his videos and they are all like that. Considering
his apparent knowledge level and how boring he is, I probably won't bother
with his other videos.
Joe Gray
W5JG
On Aug 20, 2013 11:48 AM, "Dr. David Kirkby" drkirkby@gmail.com wrote:
On 20 August 2013 15:27, Rob Klein rob.klein@smalldesign.nl wrote:
Good video for when I can't get to sleep, though ....
Yes, it does really drag on. He could say what he has to say in 25% of
the time, by cutting out all the irrelevant rubbish.
Dave
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
If you count each calibration procedure in the manual, you get 16. If you
count each scale in each procedure, it is several times that. I suspect
that the calnum is only incremented for each procedure.
Joe Gray
W5JG
On Aug 20, 2013 1:10 PM, "Joseph Gray" jgray@zianet.com wrote:
I've watched a few of his videos and they are all like that. Considering
his apparent knowledge level and how boring he is, I probably won't bother
with his other videos.
Joe Gray
W5JG
On Aug 20, 2013 11:48 AM, "Dr. David Kirkby" drkirkby@gmail.com wrote:
On 20 August 2013 15:27, Rob Klein rob.klein@smalldesign.nl wrote:
Good video for when I can't get to sleep, though ....
Yes, it does really drag on. He could say what he has to say in 25% of
the time, by cutting out all the irrelevant rubbish.
Dave
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
On 20 August 2013 17:14, Joseph Gray jgray@zianet.com wrote:
I have replaced this type of battery in other equipment, so it isn't a big
deal for me. I'm sure I can do it without losing the cal data.
Joe Gray
W5JG
Given the discussion on here a week or two ago about Agilent not
adjusting something if it is spec, there may be some advantage in not
having the ram contents kept. That way the meter is sure to require
adjustment and do wll be adjusted as close as possible. Whilst that
may not suite a metrologist who wants to track drift over time, it
suits me better to have it made as close as possible, ratther than
just left if within spec.
Dave
I've tried this on a 3458A and lost the data in the CALRAM due (I think) to
trying to read the data while the chip was still 'hot' from the removal
process. Two other 3458A's were successful.
I've tried this on a 3478A (much more like the 3457A) and lost the data and
had to recalibrate the unit. I have not sent it to Agilent for calibration
to see how 'successful' my 'home cal' was.
I know safe removal and replacement of the battery can be done and has been
done. However, I really missed knowing the 'in tolerance' status of the
3458A. I returned it to Agilent earlier this year and it was 'in tolerance'
from the calibration the previous year, so that is some consolation. The
other two 3458A's, that underwent the 'successful' CALRAM replacement, were
both 'in tolerance' when returned to Agilent and again were earlier this
year, thus giving confidence that they are stable for a long term and likely
not needing 'calibration' on a yearly basis.
I recently sent two Fluke 289, 4 1/2 digit handheld, DMM's to Fluke for
calibration, one with a Calibration Date of 08/07/09 and Calibration Counter
of 1 and the other with a Calibration Date of 02/10/11 and Calibration
Counter of 1, and both returned with 'As Found - As Left' data and no
adjustments made (Fluke Z540 Cal). This gives me great confidence in the
DMM's and their ability to be accurate for many years to come with little
need to return them to Fluke for 'cal'. It would appear that they have been
'in tolerance' since their original calibrations some 2 to 4 years ago.
From my perspective, as a 'neophyte volt-nut', knowing the 'in tolerance'
status of the DMM is very 'confidence building'. That's all I'm trying to
say.
Again, good luck.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Joseph Gray
Sent: Tuesday, August 20, 2013 11:15 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP 3457A calibration?
I have replaced this type of battery in other equipment, so it isn't a big
deal for me. I'm sure I can do it without losing the cal data.
Joe Gray
W5JG
On Aug 20, 2013 9:43 AM, "Dr. David Kirkby" drkirkby@gmail.com wrote:
On 20 August 2013 14:51, J. L. Trantham jltran@att.net wrote:
If the battery is 'OK', I would favor just sending it to Agilent for
calibration, unless you can 'SAFELY' change the battery while
preserving
the
cal data.
Good luck.
Joe
If one accepts there is some risk changing the battery while
preserving the data, one has two choices.
To me at least, the path of least risk is the one to take.
I suspect if one put a power supply in parallel with the battery with
a 10 k resistor in series, then the 10 k resistor would not drop
enough voltage to lose the ram data, but if it was accidently shorted
the current would be limited to a few hundred microamps. That's
unlikely to do any damage
I guess for a volt nut, preserving the cal data is more important than
it is to me. Not damaging the meter is higher on my priority list.
Dave
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.