Hi Everyone,
I have a problem. I bought a Solartron 7071 on Ebay.
That Voltmeter seems working properly but I think it's too noisy.
It passes Self Test and Initialize and the reading is consistent against
a Weston Cell to the µV.
But, when left with a short circuit as input the noise (maximum during 5
min) is:
0.2µV for 0.1V scale
0.4µV for 1V scale
8µV for 10V scale
Is it acceptable?
Thanks for your help.
Bye,
Jean-Louis
Solartron 7071/7081 manuals don't have official information about the
measurement noise. So I don't understand where "0.1/0.6/4 uV" is coming from.
Only one of the docs - Instruments Test Specification - have a 10 V noise limits:
"10 V noise - The rms noise of 10 readings at 10 V is measured. It must be less
than 3 uV". This is roughly equal to 12 uV p-p (with p=0.998, f=10).
My old measurements (7.5-digits, short, 5-10 minutes):
0.1 V = 0.2 uV
10 V = 5 uV
Solartron 7071/7081 have a several main sources of short-term instability. From
most important to least:
Regards,
Mickle T.
Tuesday, July 22, 2014, 8:19:25 PM, you wrote:
JLN> Hi Everyone,
JLN> I have a problem. I bought a Solartron 7071 on Ebay.
JLN> That Voltmeter seems working properly but I think it's too noisy.
JLN> It passes Self Test and Initialize and the reading is consistent against
JLN> a Weston Cell to the µV.
JLN> But, when left with a short circuit as input the noise (maximum during 5
JLN> min) is:
JLN> 0.2µV for 0.1V scale
JLN> 0.4µV for 1V scale
JLN> 8µV for 10V scale
JLN> Is it acceptable?
JLN> Thanks for your help.
JLN> Bye,
JLN> Jean-Louis
JLN> _______________________________________________
JLN> volt-nuts mailing list -- volt-nuts@febo.com
JLN> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
JLN> and follow the instructions there.
Hi Mickle,
From: "Михаил (Mickle)" timka2k@yandex.ru
My old measurements (7.5-digits, short, 5-10 minutes):
0.1 V = 0.2 uV
10 V = 5 uV
JLN> 0.2µV for 0.1V scale
JLN> 0.4µV for 1V scale
JLN> 8µV for 10V scale
So, mine seems a bit noisy!
I asked that question because if you press "DIG FILT"
it displays to the 10th of fV a bit useless if the short term stability is
not better than 0.2µV!
The web site relating your modification of the Solartron is of line.
I can provide you with a web space if you wish.
Have a good day.
Bye,
Jean-Louis