Re: [volt-nuts] Traveling Standards - Measuring Protocol

W
WarrenS
Wed, Sep 7, 2011 4:36 PM

Bob

First I'd like to say that what your doing is a great idea and could be a
great benefit to the Volt nut group.
Now for some comments that I hope you will find useful to get the most from
this exercise.

I'm still hoping you will post more details and if need be get suggestions
for the measurement that YOU should make before sending the TS out.

When doing short term stability testing better to do it at a Higher
resolution than 1.7 PPM / count.
One way to do that is to put your 6 digit voltmeter on the 1V or better yet,
the 0.1V range and measure the difference between the Device you're testing
and
a similar stable non-effected voltage such as a second device of the same
type that is not being subject to the things you're testing for.
Then taking into consideration that noise measurements are the RMS sum of
the two devices and Drift measurements are the Difference between devices,
it is possible to greatly increase the accuracy of the measurements.

Also very useful to measure the actual Temperature Coefficient of the
devices so that the temperature readings have more meaning.

When evaluation precision voltage references,
I find it very helpful to make a High Resolution strip chart recording, be
it Analog or Digital,
Preferable showing Both temperature and voltage.
For an example of how I get much better than 0.1 PPM resolution using 50
year old technology,
See attached plot comparing two 10 volt devices, using a slow, dual channel
analog Rustrak recorder and Fluke 845A/B null meter

Just my two cents worth to the bigger picture.
It is hard to properly evaluate a precision Voltage source unless you first
know the measurement uncertainties 'aka noise' including any Temperature
Coef.
AND if the temperature coef is known and not zero (over the measurement
temperatures), why not add a simple compensating voltage to reduce the
effect of temperature to insignificant levels?

ws



[volt-nuts] Traveling Standards - Measuring Protocol
Bob Smither smither at c-c-i.com

I guess it makes sense to define the set of measurements to be made on a
Traveling Standard and how those measurements are to be made.  I have
put a draft protocol for making TS measurements here:
http://c-c-i.com/tsp
This is all pretty new to me, so I would appreciate any feedback from
the list members, especially those that have agreed to make measurements.

Thanks!

Bob Smither, PhD                                Circuit Concepts, Inc.


---=====

Bob First I'd like to say that what your doing is a great idea and could be a great benefit to the Volt nut group. Now for some comments that I hope you will find useful to get the most from this exercise. I'm still hoping you will post more details and if need be get suggestions for the measurement that YOU should make before sending the TS out. When doing short term stability testing better to do it at a Higher resolution than 1.7 PPM / count. One way to do that is to put your 6 digit voltmeter on the 1V or better yet, the 0.1V range and measure the difference between the Device you're testing and a similar stable non-effected voltage such as a second device of the same type that is not being subject to the things you're testing for. Then taking into consideration that noise measurements are the RMS sum of the two devices and Drift measurements are the Difference between devices, it is possible to greatly increase the accuracy of the measurements. Also very useful to measure the actual Temperature Coefficient of the devices so that the temperature readings have more meaning. When evaluation precision voltage references, I find it very helpful to make a High Resolution strip chart recording, be it Analog or Digital, Preferable showing Both temperature and voltage. For an example of how I get much better than 0.1 PPM resolution using 50 year old technology, See attached plot comparing two 10 volt devices, using a slow, dual channel analog Rustrak recorder and Fluke 845A/B null meter Just my two cents worth to the bigger picture. It is hard to properly evaluate a precision Voltage source unless you first know the measurement uncertainties 'aka noise' including any Temperature Coef. AND if the temperature coef is known and not zero (over the measurement temperatures), why not add a simple compensating voltage to reduce the effect of temperature to insignificant levels? ws ****************** ****************** [volt-nuts] Traveling Standards - Measuring Protocol Bob Smither smither at c-c-i.com I guess it makes sense to define the set of measurements to be made on a Traveling Standard and how those measurements are to be made. I have put a draft protocol for making TS measurements here: http://c-c-i.com/tsp This is all pretty new to me, so I would appreciate any feedback from the list members, especially those that have agreed to make measurements. Thanks! -- Bob Smither, PhD Circuit Concepts, Inc. =======================================================================
AJ
Andreas Jahn
Wed, Sep 7, 2011 5:45 PM

Also very useful to measure the actual Temperature Coefficient of the
devices so that the temperature readings have more meaning.

When evaluation precision voltage references,
I find it very helpful to make a High Resolution strip chart recording, be
it Analog or Digital,
Preferable showing Both temperature and voltage.
For an example of how I get much better than 0.1 PPM resolution using 50
year old technology,
See attached plot comparing two 10 volt devices, using a slow, dual
channel
analog Rustrak recorder and Fluke 845A/B null meter

Tempco measurement will be essential for the MAX6350.
For the LM199 I have made the experience that over a
10 to 40 degree (celsius) temperature range there will be
around 1 ppm of change. Of course its better to know
the exact values. So its easier to trace back the current measurement.

First picture MAX6250A (which is similar to the MAX6350)
used as voltage reference of a LTC2400 measuring
my LM399 #2 with a LTC1043 based precision voltage divider
over temperature range 10 - 40 degree celsius.
Y-axis is 2:1 divided voltage in mV.
You can clearly see the hysteresis of the plastic device.
And you can see a "hysteresis jump" near 39 degrees of around 1 ppm.

Attached you see a plot of my LM399 #2 over temperature
as difference to my LM399 #1 at room temperature.

recorded is
red: difference voltage in mV LM399 #1 - LM399 #2
light blue: same but averaged to get the noise out
green: environment temperature of LM399 #2 housing
dark blue: within LM399 #2 housing
yellow: environment temperature of LM399 #1

so the change is around 7uV when regarding the averaged value.
and most of the "tempco" seems to be off the temperature
gradient (thermo voltages) and not from the actual temperature.

With best regards

Andreas

> Also very useful to measure the actual Temperature Coefficient of the > devices so that the temperature readings have more meaning. > > When evaluation precision voltage references, > I find it very helpful to make a High Resolution strip chart recording, be > it Analog or Digital, > Preferable showing Both temperature and voltage. > For an example of how I get much better than 0.1 PPM resolution using 50 > year old technology, > See attached plot comparing two 10 volt devices, using a slow, dual > channel > analog Rustrak recorder and Fluke 845A/B null meter > Tempco measurement will be essential for the MAX6350. For the LM199 I have made the experience that over a 10 to 40 degree (celsius) temperature range there will be around 1 ppm of change. Of course its better to know the exact values. So its easier to trace back the current measurement. First picture MAX6250A (which is similar to the MAX6350) used as voltage reference of a LTC2400 measuring my LM399 #2 with a LTC1043 based precision voltage divider over temperature range 10 - 40 degree celsius. Y-axis is 2:1 divided voltage in mV. You can clearly see the hysteresis of the plastic device. And you can see a "hysteresis jump" near 39 degrees of around 1 ppm. Attached you see a plot of my LM399 #2 over temperature as difference to my LM399 #1 at room temperature. recorded is red: difference voltage in mV LM399 #1 - LM399 #2 light blue: same but averaged to get the noise out green: environment temperature of LM399 #2 housing dark blue: within LM399 #2 housing yellow: environment temperature of LM399 #1 so the change is around 7uV when regarding the averaged value. and most of the "tempco" seems to be off the temperature gradient (thermo voltages) and not from the actual temperature. With best regards Andreas
BS
Bob Smither
Wed, Sep 7, 2011 7:12 PM

WarrenS wrote:

Bob

First I'd like to say that what your doing is a great idea and could be
a great benefit to the Volt nut group.
Now for some comments that I hope you will find useful to get the most
from this exercise.

I'm still hoping you will post more details and if need be get
suggestions for the measurement that YOU should make before sending the
TS out.

Good point - I am very open to suggestions about what I should do to
characterize the TS before it is shipped around.

When doing short term stability testing better to do it at a Higher
resolution than 1.7 PPM / count.
One way to do that is to put your 6 digit voltmeter on the 1V or better
yet,
the 0.1V range and measure the difference between the Device you're
testing and
a similar stable non-effected voltage such as a second device of the
same type that is not being subject to the things you're testing for.
Then taking into consideration that noise measurements are the RMS sum
of the two devices and Drift measurements are the Difference between
devices,
it is possible to greatly increase the accuracy of the measurements.

Also very useful to measure the actual Temperature Coefficient of the
devices so that the temperature readings have more meaning.

When evaluation precision voltage references,
I find it very helpful to make a High Resolution strip chart recording,
be it Analog or Digital,
Preferable showing Both temperature and voltage.
For an example of how I get much better than 0.1 PPM resolution using 50
year old technology,
See attached plot comparing two 10 volt devices, using a slow, dual
channel analog Rustrak recorder and Fluke 845A/B null meter

Just my two cents worth to the bigger picture.
It is hard to properly evaluate a precision Voltage source unless you
first know the measurement uncertainties 'aka noise' including any
Temperature Coef.
AND if the temperature coef is known and not zero (over the measurement
temperatures), why not add a simple compensating voltage to reduce the
effect of temperature to insignificant levels?

Hi Warren,

Thanks for the great suggestions!  I will follow up on them as I have
time.  Looks like my plate is pretty full on this project:

  1. Repackage it to avoid being visited by the Gestapo :-).
  2. Wire up a second LM199AH for the offset measurement.
  3. Measure the TC of both LM199AHs using difference technique.
  4. Measure the TC of the MAX6350 using difference technique.

I will be working towards the above.  In the meantime I will post some
very preliminary TC measurements (not high resolution) when they are
completed.

Best regards,

Bob Smither

WarrenS wrote: > Bob > > First I'd like to say that what your doing is a great idea and could be > a great benefit to the Volt nut group. > Now for some comments that I hope you will find useful to get the most > from this exercise. > > I'm still hoping you will post more details and if need be get > suggestions for the measurement that YOU should make before sending the > TS out. Good point - I am very open to suggestions about what I should do to characterize the TS before it is shipped around. > When doing short term stability testing better to do it at a Higher > resolution than 1.7 PPM / count. > One way to do that is to put your 6 digit voltmeter on the 1V or better > yet, > the 0.1V range and measure the difference between the Device you're > testing and > a similar stable non-effected voltage such as a second device of the > same type that is not being subject to the things you're testing for. > Then taking into consideration that noise measurements are the RMS sum > of the two devices and Drift measurements are the Difference between > devices, > it is possible to greatly increase the accuracy of the measurements. > > Also very useful to measure the actual Temperature Coefficient of the > devices so that the temperature readings have more meaning. > > When evaluation precision voltage references, > I find it very helpful to make a High Resolution strip chart recording, > be it Analog or Digital, > Preferable showing Both temperature and voltage. > For an example of how I get much better than 0.1 PPM resolution using 50 > year old technology, > See attached plot comparing two 10 volt devices, using a slow, dual > channel analog Rustrak recorder and Fluke 845A/B null meter > > Just my two cents worth to the bigger picture. > It is hard to properly evaluate a precision Voltage source unless you > first know the measurement uncertainties 'aka noise' including any > Temperature Coef. > AND if the temperature coef is known and not zero (over the measurement > temperatures), why not add a simple compensating voltage to reduce the > effect of temperature to insignificant levels? Hi Warren, Thanks for the great suggestions! I will follow up on them as I have time. Looks like my plate is pretty full on this project: 1. Repackage it to avoid being visited by the Gestapo :-). 2. Wire up a second LM199AH for the offset measurement. 3. Measure the TC of both LM199AHs using difference technique. 4. Measure the TC of the MAX6350 using difference technique. I will be working towards the above. In the meantime I will post some very preliminary TC measurements (not high resolution) when they are completed. Best regards, Bob Smither
MV
Mitch Van Ochten
Thu, Sep 8, 2011 12:48 AM

What type of averaging did you do for the light blue trace?

Regards,

mitch

----- Original Message -----
From: "Andreas Jahn" Andreas_-_Jahn@t-online.de
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Wednesday, September 07, 2011 1:45 PM
Subject: Re: [volt-nuts] Traveling Standards - Measuring Protocol

Also very useful to measure the actual Temperature Coefficient of the
devices so that the temperature readings have more meaning.

When evaluation precision voltage references,
I find it very helpful to make a High Resolution strip chart recording,
be
it Analog or Digital,
Preferable showing Both temperature and voltage.
For an example of how I get much better than 0.1 PPM resolution using 50
year old technology,
See attached plot comparing two 10 volt devices, using a slow, dual
channel
analog Rustrak recorder and Fluke 845A/B null meter

Tempco measurement will be essential for the MAX6350.
For the LM199 I have made the experience that over a
10 to 40 degree (celsius) temperature range there will be
around 1 ppm of change. Of course its better to know
the exact values. So its easier to trace back the current measurement.

First picture MAX6250A (which is similar to the MAX6350)
used as voltage reference of a LTC2400 measuring
my LM399 #2 with a LTC1043 based precision voltage divider
over temperature range 10 - 40 degree celsius.
Y-axis is 2:1 divided voltage in mV.
You can clearly see the hysteresis of the plastic device.
And you can see a "hysteresis jump" near 39 degrees of around 1 ppm.

Attached you see a plot of my LM399 #2 over temperature
as difference to my LM399 #1 at room temperature.

recorded is
red: difference voltage in mV LM399 #1 - LM399 #2
light blue: same but averaged to get the noise out
green: environment temperature of LM399 #2 housing
dark blue: within LM399 #2 housing
yellow: environment temperature of LM399 #1

so the change is around 7uV when regarding the averaged value.
and most of the "tempco" seems to be off the temperature
gradient (thermo voltages) and not from the actual temperature.

With best regards

Andreas



volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

What type of averaging did you do for the light blue trace? Regards, mitch ----- Original Message ----- From: "Andreas Jahn" <Andreas_-_Jahn@t-online.de> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Wednesday, September 07, 2011 1:45 PM Subject: Re: [volt-nuts] Traveling Standards - Measuring Protocol >> Also very useful to measure the actual Temperature Coefficient of the >> devices so that the temperature readings have more meaning. >> >> When evaluation precision voltage references, >> I find it very helpful to make a High Resolution strip chart recording, >> be >> it Analog or Digital, >> Preferable showing Both temperature and voltage. >> For an example of how I get much better than 0.1 PPM resolution using 50 >> year old technology, >> See attached plot comparing two 10 volt devices, using a slow, dual >> channel >> analog Rustrak recorder and Fluke 845A/B null meter >> > > Tempco measurement will be essential for the MAX6350. > For the LM199 I have made the experience that over a > 10 to 40 degree (celsius) temperature range there will be > around 1 ppm of change. Of course its better to know > the exact values. So its easier to trace back the current measurement. > > > First picture MAX6250A (which is similar to the MAX6350) > used as voltage reference of a LTC2400 measuring > my LM399 #2 with a LTC1043 based precision voltage divider > over temperature range 10 - 40 degree celsius. > Y-axis is 2:1 divided voltage in mV. > You can clearly see the hysteresis of the plastic device. > And you can see a "hysteresis jump" near 39 degrees of around 1 ppm. > > Attached you see a plot of my LM399 #2 over temperature > as difference to my LM399 #1 at room temperature. > > recorded is > red: difference voltage in mV LM399 #1 - LM399 #2 > light blue: same but averaged to get the noise out > green: environment temperature of LM399 #2 housing > dark blue: within LM399 #2 housing > yellow: environment temperature of LM399 #1 > > so the change is around 7uV when regarding the averaged value. > and most of the "tempco" seems to be off the temperature > gradient (thermo voltages) and not from the actual temperature. > > > With best regards > > Andreas > -------------------------------------------------------------------------------- > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
AJ
Andreas Jahn
Thu, Sep 8, 2011 1:56 AM

----- Original Message -----
From: "Mitch Van Ochten" mitch@vincentelectronics.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Thursday, September 08, 2011 2:48 AM
Subject: Re: [volt-nuts] Traveling Standards - Measuring Protocol

What type of averaging did you do for the light blue trace?

Regards,

mitch

----- Original Message -----
From: "Andreas Jahn" Andreas_-_Jahn@t-online.de
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Wednesday, September 07, 2011 1:45 PM
Subject: Re: [volt-nuts] Traveling Standards - Measuring Protocol

Attached you see a plot of my LM399 #2 over temperature
as difference to my LM399 #1 at room temperature.

recorded is
red: difference voltage in mV LM399 #1 - LM399 #2
light blue: same but averaged to get the noise out
green: environment temperature of LM399 #2 housing
dark blue: within LM399 #2 housing
yellow: environment temperature of LM399 #1

so the change is around 7uV when regarding the averaged value.
and most of the "tempco" seems to be off the temperature
gradient (thermo voltages) and not from the actual temperature.

Hello,

the red curve is one measurement with an integration time of 1 minute.
(about 350 measurement values of LTC2400 ADC with 10-30uVpp
noise averaged giving around 2 uVpp for 1 minute integration time).

the light blue trace is the sliding average over 21 minutes of
the red curve.10 minutes before, current minute and 10 minutes after.
(so noise is reduced down to below 0.5uVpp).

Total measuremen time (x-Axis) is around 940 minutes.

with best regards

Andreas

----- Original Message ----- From: "Mitch Van Ochten" <mitch@vincentelectronics.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Thursday, September 08, 2011 2:48 AM Subject: Re: [volt-nuts] Traveling Standards - Measuring Protocol > What type of averaging did you do for the light blue trace? > > > Regards, > > mitch > > ----- Original Message ----- > From: "Andreas Jahn" <Andreas_-_Jahn@t-online.de> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Sent: Wednesday, September 07, 2011 1:45 PM > Subject: Re: [volt-nuts] Traveling Standards - Measuring Protocol > > >> >> Attached you see a plot of my LM399 #2 over temperature >> as difference to my LM399 #1 at room temperature. >> >> recorded is >> red: difference voltage in mV LM399 #1 - LM399 #2 >> light blue: same but averaged to get the noise out >> green: environment temperature of LM399 #2 housing >> dark blue: within LM399 #2 housing >> yellow: environment temperature of LM399 #1 >> >> so the change is around 7uV when regarding the averaged value. >> and most of the "tempco" seems to be off the temperature >> gradient (thermo voltages) and not from the actual temperature. >> >> Hello, the red curve is one measurement with an integration time of 1 minute. (about 350 measurement values of LTC2400 ADC with 10-30uVpp noise averaged giving around 2 uVpp for 1 minute integration time). the light blue trace is the sliding average over 21 minutes of the red curve.10 minutes before, current minute and 10 minutes after. (so noise is reduced down to below 0.5uVpp). Total measuremen time (x-Axis) is around 940 minutes. with best regards Andreas
AJ
Andreas Jahn
Sat, Sep 24, 2011 9:38 PM

Hello Warren,

to come back to your zener diode selections:

  • do you have any source where the selection of zener diodes is described?
    I have read on a german page that zeners are selected by
    their "noise characteristic". And that the noise will show the long term
    behaviour.
    But its not described which kind of noise has an influence.
    Is it broad-band noise? 0.1 to 10 Hz noise? or even the lower frequency
    noise (which you have recorded on the strip chart recorder).

From the german link it is not clear wether the selection is

according to the value of the noise or the change before
and after pre-ageing.

And the bad news is that my LM399#1 which is drifting
much more than my LM399#2 has only about 50% of
0.1 to 10 Hz noise than the LM399#2.

By the way how do you make the plots?
Do you always record the difference of two similar devices to
compensate for the DC-Offset?
Has the strip chart recorder 1uV resolution or do you have
to add a pre-amplifier? (which one?).
The 10 second time constant: is it built in within the recorder
or external (low-ohmic / high ohmic resistor + what kind of capacitor)

with best regards

Andreas

----- Original Message -----
From: "WarrenS" warrensjmail-one@yahoo.com
To: volt-nuts@febo.com
Sent: Wednesday, September 07, 2011 6:36 PM
Subject: Re: [volt-nuts] Traveling Standards - Measuring Protocol

Bob

First I'd like to say that what your doing is a great idea and could be a
great benefit to the Volt nut group.
Now for some comments that I hope you will find useful to get the most
from
this exercise.

I'm still hoping you will post more details and if need be get suggestions
for the measurement that YOU should make before sending the TS out.

When doing short term stability testing better to do it at a Higher
resolution than 1.7 PPM / count.
One way to do that is to put your 6 digit voltmeter on the 1V or better
yet,
the 0.1V range and measure the difference between the Device you're
testing
and
a similar stable non-effected voltage such as a second device of the same
type that is not being subject to the things you're testing for.
Then taking into consideration that noise measurements are the RMS sum of
the two devices and Drift measurements are the Difference between devices,
it is possible to greatly increase the accuracy of the measurements.

Also very useful to measure the actual Temperature Coefficient of the
devices so that the temperature readings have more meaning.

When evaluation precision voltage references,
I find it very helpful to make a High Resolution strip chart recording, be
it Analog or Digital,
Preferable showing Both temperature and voltage.
For an example of how I get much better than 0.1 PPM resolution using 50
year old technology,
See attached plot comparing two 10 volt devices, using a slow, dual
channel
analog Rustrak recorder and Fluke 845A/B null meter

Just my two cents worth to the bigger picture.
It is hard to properly evaluate a precision Voltage source unless you
first
know the measurement uncertainties 'aka noise' including any Temperature
Coef.
AND if the temperature coef is known and not zero (over the measurement
temperatures), why not add a simple compensating voltage to reduce the
effect of temperature to insignificant levels?

ws



[volt-nuts] Traveling Standards - Measuring Protocol
Bob Smither smither at c-c-i.com

I guess it makes sense to define the set of measurements to be made on a
Traveling Standard and how those measurements are to be made.  I have
put a draft protocol for making TS measurements here:
http://c-c-i.com/tsp
This is all pretty new to me, so I would appreciate any feedback from
the list members, especially those that have agreed to make measurements.

Thanks!

Bob Smither, PhD                                Circuit Concepts, Inc.


---=====



volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hello Warren, to come back to your zener diode selections: - do you have any source where the selection of zener diodes is described? I have read on a german page that zeners are selected by their "noise characteristic". And that the noise will show the long term behaviour. But its not described which kind of noise has an influence. Is it broad-band noise? 0.1 to 10 Hz noise? or even the lower frequency noise (which you have recorded on the strip chart recorder). >From the german link it is not clear wether the selection is according to the value of the noise or the change before and after pre-ageing. And the bad news is that my LM399#1 which is drifting much more than my LM399#2 has only about 50% of 0.1 to 10 Hz noise than the LM399#2. By the way how do you make the plots? Do you always record the difference of two similar devices to compensate for the DC-Offset? Has the strip chart recorder 1uV resolution or do you have to add a pre-amplifier? (which one?). The 10 second time constant: is it built in within the recorder or external (low-ohmic / high ohmic resistor + what kind of capacitor) with best regards Andreas ----- Original Message ----- From: "WarrenS" <warrensjmail-one@yahoo.com> To: <volt-nuts@febo.com> Sent: Wednesday, September 07, 2011 6:36 PM Subject: Re: [volt-nuts] Traveling Standards - Measuring Protocol > Bob > > First I'd like to say that what your doing is a great idea and could be a > great benefit to the Volt nut group. > Now for some comments that I hope you will find useful to get the most > from > this exercise. > > I'm still hoping you will post more details and if need be get suggestions > for the measurement that YOU should make before sending the TS out. > > When doing short term stability testing better to do it at a Higher > resolution than 1.7 PPM / count. > One way to do that is to put your 6 digit voltmeter on the 1V or better > yet, > the 0.1V range and measure the difference between the Device you're > testing > and > a similar stable non-effected voltage such as a second device of the same > type that is not being subject to the things you're testing for. > Then taking into consideration that noise measurements are the RMS sum of > the two devices and Drift measurements are the Difference between devices, > it is possible to greatly increase the accuracy of the measurements. > > Also very useful to measure the actual Temperature Coefficient of the > devices so that the temperature readings have more meaning. > > When evaluation precision voltage references, > I find it very helpful to make a High Resolution strip chart recording, be > it Analog or Digital, > Preferable showing Both temperature and voltage. > For an example of how I get much better than 0.1 PPM resolution using 50 > year old technology, > See attached plot comparing two 10 volt devices, using a slow, dual > channel > analog Rustrak recorder and Fluke 845A/B null meter > > Just my two cents worth to the bigger picture. > It is hard to properly evaluate a precision Voltage source unless you > first > know the measurement uncertainties 'aka noise' including any Temperature > Coef. > AND if the temperature coef is known and not zero (over the measurement > temperatures), why not add a simple compensating voltage to reduce the > effect of temperature to insignificant levels? > > ws > > ****************** > ****************** > [volt-nuts] Traveling Standards - Measuring Protocol > Bob Smither smither at c-c-i.com > > I guess it makes sense to define the set of measurements to be made on a > Traveling Standard and how those measurements are to be made. I have > put a draft protocol for making TS measurements here: > http://c-c-i.com/tsp > This is all pretty new to me, so I would appreciate any feedback from > the list members, especially those that have agreed to make measurements. > > Thanks! > -- > Bob Smither, PhD Circuit Concepts, Inc. > ======================================================================= > -------------------------------------------------------------------------------- > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.