time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

U-blox teaser

RL
Robert LaJeunesse
Fri, Feb 26, 2021 2:34 PM

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS".

https://www.u-blox.com/en/blogs/insights/five-key-trends-gps

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS". https://www.u-blox.com/en/blogs/insights/five-key-trends-gps
JA
John Ackermann N8UR
Fri, Feb 26, 2021 9:27 PM

It's interesting that they talk about the F9 receivers offering 5 ns
absolute time accuracy.  Does anyone know of tests confirming that, and
what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS".

https://www.u-blox.com/en/blogs/insights/five-key-trends-gps


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy. Does anyone know of tests confirming that, and what sort of care was required in the setup to get there? John ---- On 2/26/21 9:34 AM, Robert LaJeunesse wrote: > FWIW. No detailed content, and a rather quick read. "Five key trends in GPS". > > https://www.u-blox.com/en/blogs/insights/five-key-trends-gps > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
BK
Bob kb8tq
Fri, Feb 26, 2021 10:26 PM

Hi

I can’t think of many antennas (multi band or single band) that claim to know their
delay to < 5 ns. Simply having a differential delay spec of < 5 ns is quite good.
Same thing with delay ripple, you see specs out to around 15 ns on a lot of antennas.
None of this is getting you to the actual total delay of the antenna. It’s a pretty good
bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m not sure that
the differential delay is taken out that way. Total delay, not taken out in any obvious
fashion ( at least that I can see). If the F9 has a built in antenna database, that’s not
mentioned in the doc’s. Any benefit from the corrections would have to be part of
post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns, but it would be part
of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that if the appropriate
one sigma / on a clear day / with the wind in the right direction sort of qualifiers are
attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com wrote:

It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy.  Does anyone know of tests confirming that, and what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS".
https://www.u-blox.com/en/blogs/insights/five-key-trends-gps


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi I can’t think of many antennas (multi band or single band) that claim to know their delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns is quite good. Same thing with delay ripple, you see specs out to around 15 ns on a lot of antennas. None of this is getting you to the actual total delay of the antenna. It’s a pretty good bet that number is a bit larger than either of these. Some of the ripple probably comes out in the standard modeling. I’m not sure that the differential delay is taken out that way. Total delay, not taken out in any obvious fashion ( at least that I can see). If the F9 has a built in antenna database, that’s not mentioned in the doc’s. Any benefit from the corrections would have to be part of post processing. No, that’s not the same as talking about the F9 it’s self doing X ns, but it would be part of any practical system trying to get close to 5 ns absolute accuracy. 5 ns *relative* accuracy between two F9’s? I probably could buy that if the appropriate one sigma / on a clear day / with the wind in the right direction sort of qualifiers are attached. Bob > On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> wrote: > > It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy. Does anyone know of tests confirming that, and what sort of care was required in the setup to get there? > > John > ---- > > On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >> FWIW. No detailed content, and a rather quick read. "Five key trends in GPS". >> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
EB
ed breya
Fri, Feb 26, 2021 10:37 PM

John, if you look back at the recent GPSDO discussion initiated by
"dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar,
linking a Ublox model here

https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf

I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I
studied it some, I replied back

"Thanks Leo, that's a great example of what I've been picturing. This is
from the data sheet, regarding the PPS output:

"After an initial phase of acquisition the time-pulse becomes
essentially jitter-free, generated coherently from the built-in
reference oscillator and guaranteeing an exact number of reference
frequency cycles between each time-pulse."

It looks like this is an implementation of what we've talked about here
before a few times, for saw-tooth error reduction. BTW I didn't call it
"1PPS" above, because it can be programmed for different rates - I don't
recall what they are.

BTW, did dandober ever sign up here?

Ed

John, if you look back at the recent GPSDO discussion initiated by "dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar, linking a Ublox model here https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I studied it some, I replied back "Thanks Leo, that's a great example of what I've been picturing. This is from the data sheet, regarding the PPS output: "After an initial phase of acquisition the time-pulse becomes essentially jitter-free, generated coherently from the built-in reference oscillator and guaranteeing an exact number of reference frequency cycles between each time-pulse." It looks like this is an implementation of what we've talked about here before a few times, for saw-tooth error reduction. BTW I didn't call it "1PPS" above, because it can be programmed for different rates - I don't recall what they are. BTW, did dandober ever sign up here? Ed
BK
Bob kb8tq
Fri, Feb 26, 2021 11:55 PM

Hi

Would not “absolute” timing be referenced to UTC? (or something similar)

Bob

On Feb 26, 2021, at 5:37 PM, ed breya eb@telight.com wrote:

John, if you look back at the recent GPSDO discussion initiated by "dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar, linking a Ublox model here

https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf

I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I studied it some, I replied back

"Thanks Leo, that's a great example of what I've been picturing. This is from the data sheet, regarding the PPS output:

"After an initial phase of acquisition the time-pulse becomes essentially jitter-free, generated coherently from the built-in reference oscillator and guaranteeing an exact number of reference frequency cycles between each time-pulse."

It looks like this is an implementation of what we've talked about here before a few times, for saw-tooth error reduction. BTW I didn't call it "1PPS" above, because it can be programmed for different rates - I don't recall what they are.

BTW, did dandober ever sign up here?

Ed


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi Would not “absolute” timing be referenced to UTC? (or something similar) Bob > On Feb 26, 2021, at 5:37 PM, ed breya <eb@telight.com> wrote: > > John, if you look back at the recent GPSDO discussion initiated by "dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar, linking a Ublox model here > > https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf > > I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I studied it some, I replied back > > "Thanks Leo, that's a great example of what I've been picturing. This is from the data sheet, regarding the PPS output: > > "After an initial phase of acquisition the time-pulse becomes essentially jitter-free, generated coherently from the built-in reference oscillator and guaranteeing an exact number of reference frequency cycles between each time-pulse." > > It looks like this is an implementation of what we've talked about here before a few times, for saw-tooth error reduction. BTW I didn't call it "1PPS" above, because it can be programmed for different rates - I don't recall what they are. > > BTW, did dandober ever sign up here? > > Ed > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
JA
John Ackermann N8UR
Sat, Feb 27, 2021 12:00 AM

They're claiming "even better than" 5 ns for relative time, which given
the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that claim to know their
delay to < 5 ns. Simply having a differential delay spec of < 5 ns is quite good.
Same thing with delay ripple, you see specs out to around 15 ns on a lot of antennas.
None of this is getting you to the actual total delay of the antenna. It’s a pretty good
bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m not sure that
the differential delay is taken out that way. Total delay, not taken out in any obvious
fashion ( at least that I can see). If the F9 has a built in antenna database, that’s not
mentioned in the doc’s. Any benefit from the corrections would have to be part of
post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns, but it would be part
of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that if the appropriate
one sigma / on a clear day / with the wind in the right direction sort of qualifiers are
attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com wrote:

It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy.  Does anyone know of tests confirming that, and what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends in GPS".
https://www.u-blox.com/en/blogs/insights/five-key-trends-gps


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

They're claiming "even better than" 5 ns for relative time, which given the 4 ns jitter seems at least sort-of reasonable. But until someone shows me otherwise, I'm still thinking that getting better than 25 ns absolute accuracy is a pretty good day's work. John ---- On 2/26/21 5:26 PM, Bob kb8tq wrote: > Hi > > I can’t think of many antennas (multi band or single band) that claim to know their > delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns is quite good. > Same thing with delay ripple, you see specs out to around 15 ns on a lot of antennas. > None of this is getting you to the actual total delay of the antenna. It’s a pretty good > bet that number is a bit larger than either of these. > > Some of the ripple probably comes out in the standard modeling. I’m not sure that > the differential delay is taken out that way. Total delay, not taken out in any obvious > fashion ( at least that I can see). If the F9 has a built in antenna database, that’s not > mentioned in the doc’s. Any benefit from the corrections would have to be part of > post processing. > > No, that’s not the same as talking about the F9 it’s self doing X ns, but it would be part > of any practical system trying to get close to 5 ns absolute accuracy. > > 5 ns *relative* accuracy between two F9’s? I probably could buy that if the appropriate > one sigma / on a clear day / with the wind in the right direction sort of qualifiers are > attached. > > Bob > >> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> wrote: >> >> It's interesting that they talk about the F9 receivers offering 5 ns absolute time accuracy. Does anyone know of tests confirming that, and what sort of care was required in the setup to get there? >> >> John >> ---- >> >> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >>> FWIW. No detailed content, and a rather quick read. "Five key trends in GPS". >>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
JA
John Ackermann N8UR
Sat, Feb 27, 2021 12:31 AM

You have to be careful comparing the LEA-M8F with other GPS units.  It
does have low jitter because the "TIMEPULSE" signal is derived from the
TCXO which is locked to the GPS time mark.

But their claim of "essentially jitter free" depends on your definition
of "essentially" -- attached is a comparison of LEA-M8F PPS jitter vs.
NEO-M8T and ZED-F9T raw, and NEO-M8T and ZED-F9T sawtooth corrected PPS.
The M8F is definitely better than the raw M8T and even the raw F9T,
but the corrected M8T and F9T are both much better than the M8F.  And
the M8F does not have sawtooth correction available.

So, it's a neat implementation and has some applications (basically as a
modest performance 30.72 MHz GPSDO), but TANSTAAFL.

John

On 2/26/21 5:37 PM, ed breya wrote:

John, if you look back at the recent GPSDO discussion initiated by
"dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar,
linking a Ublox model here

https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf

I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I
studied it some, I replied back

"Thanks Leo, that's a great example of what I've been picturing. This is
from the data sheet, regarding the PPS output:

"After an initial phase of acquisition the time-pulse becomes
essentially jitter-free, generated coherently from the built-in
reference oscillator and guaranteeing an exact number of reference
frequency cycles between each time-pulse."

It looks like this is an implementation of what we've talked about here
before a few times, for saw-tooth error reduction. BTW I didn't call it
"1PPS" above, because it can be programmed for different rates - I don't
recall what they are.

BTW, did dandober ever sign up here?

Ed


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

You have to be careful comparing the LEA-M8F with other GPS units. It does have low jitter because the "TIMEPULSE" signal is derived from the TCXO which is locked to the GPS time mark. But their claim of "essentially jitter free" depends on your definition of "essentially" -- attached is a comparison of LEA-M8F PPS jitter vs. NEO-M8T and ZED-F9T raw, and NEO-M8T and ZED-F9T sawtooth corrected PPS. The M8F is definitely better than the raw M8T and even the raw F9T, but the corrected M8T and F9T are both much better than the M8F. And the M8F does *not* have sawtooth correction available. So, it's a neat implementation and has some applications (basically as a modest performance 30.72 MHz GPSDO), but TANSTAAFL. John ---- On 2/26/21 5:37 PM, ed breya wrote: > John, if you look back at the recent GPSDO discussion initiated by > "dandober" in hpaligent keysight group, you'll see a post by Leo Bodnar, > linking a Ublox model here > > https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf > > > I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After I > studied it some, I replied back > > "Thanks Leo, that's a great example of what I've been picturing. This is > from the data sheet, regarding the PPS output: > > "After an initial phase of acquisition the time-pulse becomes > essentially jitter-free, generated coherently from the built-in > reference oscillator and guaranteeing an exact number of reference > frequency cycles between each time-pulse." > > It looks like this is an implementation of what we've talked about here > before a few times, for saw-tooth error reduction. BTW I didn't call it > "1PPS" above, because it can be programmed for different rates - I don't > recall what they are. > > BTW, did dandober ever sign up here? > > Ed > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
JA
John Ackermann
Sat, Feb 27, 2021 1:01 AM

That's how I'd interpret it.

⁣Get BlueMail for Android ​

On Feb 26, 2021, 7:42 PM, at 7:42 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

Would not “absolute” timing be referenced to UTC? (or something
similar)

Bob

On Feb 26, 2021, at 5:37 PM, ed breya eb@telight.com wrote:

John, if you look back at the recent GPSDO discussion initiated by

"dandober" in hpaligent keysight group, you'll see a post by Leo
Bodnar, linking a Ublox model here

I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After

I studied it some, I replied back

"Thanks Leo, that's a great example of what I've been picturing. This

is from the data sheet, regarding the PPS output:

"After an initial phase of acquisition the time-pulse becomes

essentially jitter-free, generated coherently from the built-in
reference oscillator and guaranteeing an exact number of reference
frequency cycles between each time-pulse."

It looks like this is an implementation of what we've talked about

here before a few times, for saw-tooth error reduction. BTW I didn't
call it "1PPS" above, because it can be programmed for different rates

  • I don't recall what they are.

BTW, did dandober ever sign up here?

Ed


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

That's how I'd interpret it. ⁣Get BlueMail for Android ​ On Feb 26, 2021, 7:42 PM, at 7:42 PM, Bob kb8tq <kb8tq@n1k.org> wrote: >Hi > >Would not “absolute” timing be referenced to UTC? (or something >similar) > >Bob > >> On Feb 26, 2021, at 5:37 PM, ed breya <eb@telight.com> wrote: >> >> John, if you look back at the recent GPSDO discussion initiated by >"dandober" in hpaligent keysight group, you'll see a post by Leo >Bodnar, linking a Ublox model here >> >> >https://www.u-blox.com/sites/default/files/products/documents/LEA-M8F_DataSheet_%28UBX-14001772%29.pdf >> >> I looked it up, and I recall the PPS timing spec is +/- 5 nSec. After >I studied it some, I replied back >> >> "Thanks Leo, that's a great example of what I've been picturing. This >is from the data sheet, regarding the PPS output: >> >> "After an initial phase of acquisition the time-pulse becomes >essentially jitter-free, generated coherently from the built-in >reference oscillator and guaranteeing an exact number of reference >frequency cycles between each time-pulse." >> >> It looks like this is an implementation of what we've talked about >here before a few times, for saw-tooth error reduction. BTW I didn't >call it "1PPS" above, because it can be programmed for different rates >- I don't recall what they are. >> >> BTW, did dandober ever sign up here? >> >> Ed >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > >_______________________________________________ >time-nuts mailing list -- time-nuts@lists.febo.com >To unsubscribe, go to >http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >and follow the instructions there.
MW
Michael Wouters
Sat, Feb 27, 2021 1:02 AM

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor and the
technique used was just to measure the electronic delay ie by injecting a
signal into the circuit. To do it properly, you need a setup in a microwave
anechoic chamber with transmitting antenna etc. The practical difference
may be small though, 1 or 2 ns ( sample of one antenna!).

Cheers
Michael

On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com wrote:

They're claiming "even better than" 5 ns for relative time, which given
the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that claim to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a lot

of antennas.

None of this is getting you to the actual total delay of the antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m not

sure that

the differential delay is taken out that way. Total delay, not taken out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns,

but it would be part

of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that if

the appropriate

one sigma / on a clear day / with the wind in the right direction sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com wrote:

It's interesting that they talk about the F9 receivers offering 5 ns

absolute time accuracy.  Does anyone know of tests confirming that, and
what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Typical L1 antenna delays range from 20 to 70 ns. I know of only one antenna for which a delay is given by the vendor and the technique used was just to measure the electronic delay ie by injecting a signal into the circuit. To do it properly, you need a setup in a microwave anechoic chamber with transmitting antenna etc. The practical difference may be small though, 1 or 2 ns ( sample of one antenna!). Cheers Michael On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> wrote: > They're claiming "even better than" 5 ns for relative time, which given > the 4 ns jitter seems at least sort-of reasonable. But until someone > shows me otherwise, I'm still thinking that getting better than 25 ns > absolute accuracy is a pretty good day's work. > > John > ---- > > On 2/26/21 5:26 PM, Bob kb8tq wrote: > > Hi > > > > I can’t think of many antennas (multi band or single band) that claim to > know their > > delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns is > quite good. > > Same thing with delay ripple, you see specs out to around 15 ns on a lot > of antennas. > > None of this is getting you to the actual total delay of the antenna. > It’s a pretty good > > bet that number is a bit larger than either of these. > > > > Some of the ripple probably comes out in the standard modeling. I’m not > sure that > > the differential delay is taken out that way. Total delay, not taken out > in any obvious > > fashion ( at least that I can see). If the F9 has a built in antenna > database, that’s not > > mentioned in the doc’s. Any benefit from the corrections would have to > be part of > > post processing. > > > > No, that’s not the same as talking about the F9 it’s self doing X ns, > but it would be part > > of any practical system trying to get close to 5 ns absolute accuracy. > > > > 5 ns *relative* accuracy between two F9’s? I probably could buy that if > the appropriate > > one sigma / on a clear day / with the wind in the right direction sort > of qualifiers are > > attached. > > > > Bob > > > >> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> wrote: > >> > >> It's interesting that they talk about the F9 receivers offering 5 ns > absolute time accuracy. Does anyone know of tests confirming that, and > what sort of care was required in the setup to get there? > >> > >> John > >> ---- > >> > >> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: > >>> FWIW. No detailed content, and a rather quick read. "Five key trends > in GPS". > >>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps > >>> _______________________________________________ > >>> time-nuts mailing list -- time-nuts@lists.febo.com > >>> To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>> and follow the instructions there. > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >> and follow the instructions there. > > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
LJ
Lux, Jim
Sat, Feb 27, 2021 1:06 AM

On 2/26/21 4:00 PM, John Ackermann N8UR wrote:

They're claiming "even better than" 5 ns for relative time, which
given the 4 ns jitter seems at least sort-of reasonable.  But until
someone shows me otherwise, I'm still thinking that getting better
than 25 ns absolute accuracy is a pretty good day's work.

John

And if the jitter is uniformly distributed (i.e. 1 clock cycle of some
sort) - then the sd is 1/sqrt(12) of the extremes, so a "max
uncertainty" of 5ns, is more like a 1.5 ns 1 sigma.

On 2/26/21 4:00 PM, John Ackermann N8UR wrote: > They're claiming "even better than" 5 ns for relative time, which > given the 4 ns jitter seems at least sort-of reasonable.  But until > someone shows me otherwise, I'm still thinking that getting better > than 25 ns absolute accuracy is a pretty good day's work. > > John > ---- And if the jitter is uniformly distributed (i.e. 1 clock cycle of some sort) - then the sd is 1/sqrt(12) of the extremes, so a "max uncertainty" of 5ns, is more like a 1.5 ns 1 sigma.
JA
John Ackermann N8UR
Sat, Feb 27, 2021 2:29 AM

A while ago I tried doing a decidedly non-anechoic measurement with a
VNA exciter going to a 1500 MHz ground plane and the receiver connected
to the antenna (with a known delay cable) and I got a similar result,
but there was enough noise that I didn't think I could nail it down to
within 10 ns.

I've also measured GPS antenna splitters and they tend to have 20-ish ns
delays, mainly due to the SAW filters.  I did surgery on an HP splitter
to remove the filters so it could be used for L1 and L2 and that dropped
the delay down to only 1 or 2 ns.

So there's definitely lots of stuff to calibrate if you want to get
accurate time transfer.

John

On 2/26/21 8:02 PM, Michael Wouters wrote:

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor and the
technique used was just to measure the electronic delay ie by injecting a
signal into the circuit. To do it properly, you need a setup in a microwave
anechoic chamber with transmitting antenna etc. The practical difference
may be small though, 1 or 2 ns ( sample of one antenna!).

Cheers
Michael

On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com wrote:

They're claiming "even better than" 5 ns for relative time, which given
the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that claim to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a lot

of antennas.

None of this is getting you to the actual total delay of the antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m not

sure that

the differential delay is taken out that way. Total delay, not taken out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns,

but it would be part

of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that if

the appropriate

one sigma / on a clear day / with the wind in the right direction sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com wrote:

It's interesting that they talk about the F9 receivers offering 5 ns

absolute time accuracy.  Does anyone know of tests confirming that, and
what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

A while ago I tried doing a decidedly non-anechoic measurement with a VNA exciter going to a 1500 MHz ground plane and the receiver connected to the antenna (with a known delay cable) and I got a similar result, but there was enough noise that I didn't think I could nail it down to within 10 ns. I've also measured GPS antenna splitters and they tend to have 20-ish ns delays, mainly due to the SAW filters. I did surgery on an HP splitter to remove the filters so it could be used for L1 and L2 and that dropped the delay down to only 1 or 2 ns. So there's definitely lots of stuff to calibrate if you want to get accurate time transfer. John ---- On 2/26/21 8:02 PM, Michael Wouters wrote: > Typical L1 antenna delays range from 20 to 70 ns. > I know of only one antenna for which a delay is given by the vendor and the > technique used was just to measure the electronic delay ie by injecting a > signal into the circuit. To do it properly, you need a setup in a microwave > anechoic chamber with transmitting antenna etc. The practical difference > may be small though, 1 or 2 ns ( sample of one antenna!). > > Cheers > Michael > > On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> wrote: > >> They're claiming "even better than" 5 ns for relative time, which given >> the 4 ns jitter seems at least sort-of reasonable. But until someone >> shows me otherwise, I'm still thinking that getting better than 25 ns >> absolute accuracy is a pretty good day's work. >> >> John >> ---- >> >> On 2/26/21 5:26 PM, Bob kb8tq wrote: >>> Hi >>> >>> I can’t think of many antennas (multi band or single band) that claim to >> know their >>> delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns is >> quite good. >>> Same thing with delay ripple, you see specs out to around 15 ns on a lot >> of antennas. >>> None of this is getting you to the actual total delay of the antenna. >> It’s a pretty good >>> bet that number is a bit larger than either of these. >>> >>> Some of the ripple probably comes out in the standard modeling. I’m not >> sure that >>> the differential delay is taken out that way. Total delay, not taken out >> in any obvious >>> fashion ( at least that I can see). If the F9 has a built in antenna >> database, that’s not >>> mentioned in the doc’s. Any benefit from the corrections would have to >> be part of >>> post processing. >>> >>> No, that’s not the same as talking about the F9 it’s self doing X ns, >> but it would be part >>> of any practical system trying to get close to 5 ns absolute accuracy. >>> >>> 5 ns *relative* accuracy between two F9’s? I probably could buy that if >> the appropriate >>> one sigma / on a clear day / with the wind in the right direction sort >> of qualifiers are >>> attached. >>> >>> Bob >>> >>>> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> wrote: >>>> >>>> It's interesting that they talk about the F9 receivers offering 5 ns >> absolute time accuracy. Does anyone know of tests confirming that, and >> what sort of care was required in the setup to get there? >>>> >>>> John >>>> ---- >>>> >>>> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >>>>> FWIW. No detailed content, and a rather quick read. "Five key trends >> in GPS". >>>>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>> and follow the instructions there. >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >>> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
BK
Bob kb8tq
Sat, Feb 27, 2021 1:39 PM

Hi

The same 20 or so ns delay in a saw would also apply to the
saw (or tight filter) in some timing antennas. It also would apply
to the saw(s) in some modules. Even if the tolerance is “only”
a couple ns on each of them, you could have 3 or more in the
chain.

Lots of numbers to crunch to get to 5 ns “absolute”. One could go
grab a GPS simulator and start poking. First step would be to find
a simulator that is spec’d for a < 5 ns tolerance on the PPS into
GPS out. I do believe that rules out the eBay marvels that some
of us have lying around …..

Simpler answer would be a quick “clock trip” with your car full
of 5071’s …… hour drive over to NIST and then back home.
That sounds practical for most of us :) :)

Bob

On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR jra@febo.com wrote:

A while ago I tried doing a decidedly non-anechoic measurement with a VNA exciter going to a 1500 MHz ground plane and the receiver connected to the antenna (with a known delay cable) and I got a similar result, but there was enough noise that I didn't think I could nail it down to within 10 ns.

I've also measured GPS antenna splitters and they tend to have 20-ish ns delays, mainly due to the SAW filters.  I did surgery on an HP splitter to remove the filters so it could be used for L1 and L2 and that dropped the delay down to only 1 or 2 ns.

So there's definitely lots of stuff to calibrate if you want to get accurate time transfer.

John

On 2/26/21 8:02 PM, Michael Wouters wrote:

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor and the
technique used was just to measure the electronic delay ie by injecting a
signal into the circuit. To do it properly, you need a setup in a microwave
anechoic chamber with transmitting antenna etc. The practical difference
may be small though, 1 or 2 ns ( sample of one antenna!).
Cheers
Michael
On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com wrote:

They're claiming "even better than" 5 ns for relative time, which given
the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that claim to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a lot

of antennas.

None of this is getting you to the actual total delay of the antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m not

sure that

the differential delay is taken out that way. Total delay, not taken out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns,

but it would be part

of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that if

the appropriate

one sigma / on a clear day / with the wind in the right direction sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com wrote:

It's interesting that they talk about the F9 receivers offering 5 ns

absolute time accuracy.  Does anyone know of tests confirming that, and
what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi The same 20 or so ns delay in a saw would also apply to the saw (or tight filter) in some timing antennas. It also would apply to the saw(s) in some modules. Even if the tolerance is “only” a couple ns on each of them, you *could* have 3 or more in the chain. Lots of numbers to crunch to get to 5 ns “absolute”. One could go grab a GPS simulator and start poking. First step would be to find a simulator that is spec’d for a < 5 ns tolerance on the PPS into GPS out. I do believe that rules out the eBay marvels that some of us have lying around ….. Simpler answer would be a quick “clock trip” with your car full of 5071’s …… hour drive over to NIST and then back home. That sounds practical for most of us :) :) Bob > On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR <jra@febo.com> wrote: > > A while ago I tried doing a decidedly non-anechoic measurement with a VNA exciter going to a 1500 MHz ground plane and the receiver connected to the antenna (with a known delay cable) and I got a similar result, but there was enough noise that I didn't think I could nail it down to within 10 ns. > > I've also measured GPS antenna splitters and they tend to have 20-ish ns delays, mainly due to the SAW filters. I did surgery on an HP splitter to remove the filters so it could be used for L1 and L2 and that dropped the delay down to only 1 or 2 ns. > > So there's definitely lots of stuff to calibrate if you want to get accurate time transfer. > > John > ---- > > On 2/26/21 8:02 PM, Michael Wouters wrote: >> Typical L1 antenna delays range from 20 to 70 ns. >> I know of only one antenna for which a delay is given by the vendor and the >> technique used was just to measure the electronic delay ie by injecting a >> signal into the circuit. To do it properly, you need a setup in a microwave >> anechoic chamber with transmitting antenna etc. The practical difference >> may be small though, 1 or 2 ns ( sample of one antenna!). >> Cheers >> Michael >> On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> wrote: >>> They're claiming "even better than" 5 ns for relative time, which given >>> the 4 ns jitter seems at least sort-of reasonable. But until someone >>> shows me otherwise, I'm still thinking that getting better than 25 ns >>> absolute accuracy is a pretty good day's work. >>> >>> John >>> ---- >>> >>> On 2/26/21 5:26 PM, Bob kb8tq wrote: >>>> Hi >>>> >>>> I can’t think of many antennas (multi band or single band) that claim to >>> know their >>>> delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns is >>> quite good. >>>> Same thing with delay ripple, you see specs out to around 15 ns on a lot >>> of antennas. >>>> None of this is getting you to the actual total delay of the antenna. >>> It’s a pretty good >>>> bet that number is a bit larger than either of these. >>>> >>>> Some of the ripple probably comes out in the standard modeling. I’m not >>> sure that >>>> the differential delay is taken out that way. Total delay, not taken out >>> in any obvious >>>> fashion ( at least that I can see). If the F9 has a built in antenna >>> database, that’s not >>>> mentioned in the doc’s. Any benefit from the corrections would have to >>> be part of >>>> post processing. >>>> >>>> No, that’s not the same as talking about the F9 it’s self doing X ns, >>> but it would be part >>>> of any practical system trying to get close to 5 ns absolute accuracy. >>>> >>>> 5 ns *relative* accuracy between two F9’s? I probably could buy that if >>> the appropriate >>>> one sigma / on a clear day / with the wind in the right direction sort >>> of qualifiers are >>>> attached. >>>> >>>> Bob >>>> >>>>> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> wrote: >>>>> >>>>> It's interesting that they talk about the F9 receivers offering 5 ns >>> absolute time accuracy. Does anyone know of tests confirming that, and >>> what sort of care was required in the setup to get there? >>>>> >>>>> John >>>>> ---- >>>>> >>>>> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >>>>>> FWIW. No detailed content, and a rather quick read. "Five key trends >>> in GPS". >>>>>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>> To unsubscribe, go to >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>> and follow the instructions there. >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>> To unsubscribe, go to >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>> and follow the instructions there. >>>> >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>> To unsubscribe, go to >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>> and follow the instructions there. >>>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
DW
Dana Whitlow
Sat, Feb 27, 2021 2:41 PM

I've long understood that ionospheric delays and variations thereof lead to
position
uncertainties in GPS navigation receivers, to the tune of perhaps 10m
(2DRMS IIRC).,
and that these are said to constitute the single largest GPS error source.

Q1: Would this not imply timing errors of comparable magnitude (10's of
nsec)
for a single band GPS?

Q2: Why have I not seen this issue raised in connection with the present
discussion
about achievable absolute timing accuracy?

Q3: Do any of the "modern" timing GPS receivers available to civilians do
dual-band
reception in a way that includes estimation of (and correction for)
said delays and
their variations?  I know that Garmin, for one, is now selling L1/L5
handheld GPS
receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication
that these
units make any attempt at doing such corrections.

Dana

On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The same 20 or so ns delay in a saw would also apply to the
saw (or tight filter) in some timing antennas. It also would apply
to the saw(s) in some modules. Even if the tolerance is “only”
a couple ns on each of them, you could have 3 or more in the
chain.

Lots of numbers to crunch to get to 5 ns “absolute”. One could go
grab a GPS simulator and start poking. First step would be to find
a simulator that is spec’d for a < 5 ns tolerance on the PPS into
GPS out. I do believe that rules out the eBay marvels that some
of us have lying around …..

Simpler answer would be a quick “clock trip” with your car full
of 5071’s …… hour drive over to NIST and then back home.
That sounds practical for most of us :) :)

Bob

On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR jra@febo.com wrote:

A while ago I tried doing a decidedly non-anechoic measurement with a

VNA exciter going to a 1500 MHz ground plane and the receiver connected to
the antenna (with a known delay cable) and I got a similar result, but
there was enough noise that I didn't think I could nail it down to within
10 ns.

I've also measured GPS antenna splitters and they tend to have 20-ish ns

delays, mainly due to the SAW filters.  I did surgery on an HP splitter to
remove the filters so it could be used for L1 and L2 and that dropped the
delay down to only 1 or 2 ns.

So there's definitely lots of stuff to calibrate if you want to get

accurate time transfer.

John

On 2/26/21 8:02 PM, Michael Wouters wrote:

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor and

the

technique used was just to measure the electronic delay ie by injecting

a

signal into the circuit. To do it properly, you need a setup in a

microwave

anechoic chamber with transmitting antenna etc. The practical difference
may be small though, 1 or 2 ns ( sample of one antenna!).
Cheers
Michael
On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com

wrote:

They're claiming "even better than" 5 ns for relative time, which given
the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that claim

to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns

is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a

lot

of antennas.

None of this is getting you to the actual total delay of the antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m

not

sure that

the differential delay is taken out that way. Total delay, not taken

out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns,

but it would be part

of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that

if

the appropriate

one sigma / on a clear day / with the wind in the right direction sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com

wrote:

It's interesting that they talk about the F9 receivers offering 5 ns

absolute time accuracy.  Does anyone know of tests confirming that, and
what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

I've long understood that ionospheric delays and variations thereof lead to *position* uncertainties in GPS navigation receivers, to the tune of perhaps 10m (2DRMS IIRC)., and that these are said to constitute the single largest GPS error source. Q1: Would this not imply timing errors of comparable magnitude (10's of nsec) for a single band GPS? Q2: Why have I not seen this issue raised in connection with the present discussion about achievable absolute timing accuracy? Q3: Do any of the "modern" timing GPS receivers available to civilians do dual-band reception in a way that includes estimation of (and correction for) said delays and their variations? I know that Garmin, for one, is now selling L1/L5 handheld GPS receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication that these units make any attempt at doing such corrections. Dana On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > The same 20 or so ns delay in a saw would also apply to the > saw (or tight filter) in some timing antennas. It also would apply > to the saw(s) in some modules. Even if the tolerance is “only” > a couple ns on each of them, you *could* have 3 or more in the > chain. > > Lots of numbers to crunch to get to 5 ns “absolute”. One could go > grab a GPS simulator and start poking. First step would be to find > a simulator that is spec’d for a < 5 ns tolerance on the PPS into > GPS out. I do believe that rules out the eBay marvels that some > of us have lying around ….. > > Simpler answer would be a quick “clock trip” with your car full > of 5071’s …… hour drive over to NIST and then back home. > That sounds practical for most of us :) :) > > Bob > > > On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR <jra@febo.com> wrote: > > > > A while ago I tried doing a decidedly non-anechoic measurement with a > VNA exciter going to a 1500 MHz ground plane and the receiver connected to > the antenna (with a known delay cable) and I got a similar result, but > there was enough noise that I didn't think I could nail it down to within > 10 ns. > > > > I've also measured GPS antenna splitters and they tend to have 20-ish ns > delays, mainly due to the SAW filters. I did surgery on an HP splitter to > remove the filters so it could be used for L1 and L2 and that dropped the > delay down to only 1 or 2 ns. > > > > So there's definitely lots of stuff to calibrate if you want to get > accurate time transfer. > > > > John > > ---- > > > > On 2/26/21 8:02 PM, Michael Wouters wrote: > >> Typical L1 antenna delays range from 20 to 70 ns. > >> I know of only one antenna for which a delay is given by the vendor and > the > >> technique used was just to measure the electronic delay ie by injecting > a > >> signal into the circuit. To do it properly, you need a setup in a > microwave > >> anechoic chamber with transmitting antenna etc. The practical difference > >> may be small though, 1 or 2 ns ( sample of one antenna!). > >> Cheers > >> Michael > >> On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> > wrote: > >>> They're claiming "even better than" 5 ns for relative time, which given > >>> the 4 ns jitter seems at least sort-of reasonable. But until someone > >>> shows me otherwise, I'm still thinking that getting better than 25 ns > >>> absolute accuracy is a pretty good day's work. > >>> > >>> John > >>> ---- > >>> > >>> On 2/26/21 5:26 PM, Bob kb8tq wrote: > >>>> Hi > >>>> > >>>> I can’t think of many antennas (multi band or single band) that claim > to > >>> know their > >>>> delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns > is > >>> quite good. > >>>> Same thing with delay ripple, you see specs out to around 15 ns on a > lot > >>> of antennas. > >>>> None of this is getting you to the actual total delay of the antenna. > >>> It’s a pretty good > >>>> bet that number is a bit larger than either of these. > >>>> > >>>> Some of the ripple probably comes out in the standard modeling. I’m > not > >>> sure that > >>>> the differential delay is taken out that way. Total delay, not taken > out > >>> in any obvious > >>>> fashion ( at least that I can see). If the F9 has a built in antenna > >>> database, that’s not > >>>> mentioned in the doc’s. Any benefit from the corrections would have to > >>> be part of > >>>> post processing. > >>>> > >>>> No, that’s not the same as talking about the F9 it’s self doing X ns, > >>> but it would be part > >>>> of any practical system trying to get close to 5 ns absolute accuracy. > >>>> > >>>> 5 ns *relative* accuracy between two F9’s? I probably could buy that > if > >>> the appropriate > >>>> one sigma / on a clear day / with the wind in the right direction sort > >>> of qualifiers are > >>>> attached. > >>>> > >>>> Bob > >>>> > >>>>> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> > wrote: > >>>>> > >>>>> It's interesting that they talk about the F9 receivers offering 5 ns > >>> absolute time accuracy. Does anyone know of tests confirming that, and > >>> what sort of care was required in the setup to get there? > >>>>> > >>>>> John > >>>>> ---- > >>>>> > >>>>> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: > >>>>>> FWIW. No detailed content, and a rather quick read. "Five key trends > >>> in GPS". > >>>>>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps > >>>>>> _______________________________________________ > >>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>> To unsubscribe, go to > >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>> and follow the instructions there. > >>>>> > >>>>> _______________________________________________ > >>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>> To unsubscribe, go to > >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>> and follow the instructions there. > >>>> > >>>> > >>>> _______________________________________________ > >>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>> To unsubscribe, go to > >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>> and follow the instructions there. > >>>> > >>> > >>> _______________________________________________ > >>> time-nuts mailing list -- time-nuts@lists.febo.com > >>> To unsubscribe, go to > >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>> and follow the instructions there. > >>> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >> and follow the instructions there. > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
BK
Bob kb8tq
Sat, Feb 27, 2021 3:05 PM

Hi

On Feb 27, 2021, at 9:41 AM, Dana Whitlow k8yumdoober@gmail.com wrote:

I've long understood that ionospheric delays and variations thereof lead to
position
uncertainties in GPS navigation receivers, to the tune of perhaps 10m
(2DRMS IIRC).,
and that these are said to constitute the single largest GPS error source.

Q1: Would this not imply timing errors of comparable magnitude (10's of
nsec)
for a single band GPS?

Once all the signals “hit” the antenna, the delays are mostly common mode.
Instead of showing up as a position error, they show up as an error in the
time estimate. Since time is one of the things you estimate in the solution
(along with X,Y, and Z) it get’s it’s own independent solution.

Q2: Why have I not seen this issue raised in connection with the present
discussion
about achievable absolute timing accuracy?

GPS time transfer is often done to the sub-ns level. There are a number of
papers on this.

Q3: Do any of the "modern" timing GPS receivers available to civilians do
dual-band
reception in a way that includes estimation of (and correction for)
said delays and
their variations?  I know that Garmin, for one, is now selling L1/L5
handheld GPS
receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication
that these
units make any attempt at doing such corrections.

Yes, some receivers do an estimate of ionospheric delay based on the
variation of that delay with frequency. This does not help with tropospheric
delay or all of the various “common mode” issues we have been talking about.
It is also unclear how the “unknown” timing variation between the bands
due to the antenna impacts these solutions…..

Bob

Dana

On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The same 20 or so ns delay in a saw would also apply to the
saw (or tight filter) in some timing antennas. It also would apply
to the saw(s) in some modules. Even if the tolerance is “only”
a couple ns on each of them, you could have 3 or more in the
chain.

Lots of numbers to crunch to get to 5 ns “absolute”. One could go
grab a GPS simulator and start poking. First step would be to find
a simulator that is spec’d for a < 5 ns tolerance on the PPS into
GPS out. I do believe that rules out the eBay marvels that some
of us have lying around …..

Simpler answer would be a quick “clock trip” with your car full
of 5071’s …… hour drive over to NIST and then back home.
That sounds practical for most of us :) :)

Bob

On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR jra@febo.com wrote:

A while ago I tried doing a decidedly non-anechoic measurement with a

VNA exciter going to a 1500 MHz ground plane and the receiver connected to
the antenna (with a known delay cable) and I got a similar result, but
there was enough noise that I didn't think I could nail it down to within
10 ns.

I've also measured GPS antenna splitters and they tend to have 20-ish ns

delays, mainly due to the SAW filters.  I did surgery on an HP splitter to
remove the filters so it could be used for L1 and L2 and that dropped the
delay down to only 1 or 2 ns.

So there's definitely lots of stuff to calibrate if you want to get

accurate time transfer.

John

On 2/26/21 8:02 PM, Michael Wouters wrote:

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor and

the

technique used was just to measure the electronic delay ie by injecting

a

signal into the circuit. To do it properly, you need a setup in a

microwave

anechoic chamber with transmitting antenna etc. The practical difference
may be small though, 1 or 2 ns ( sample of one antenna!).
Cheers
Michael
On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com

wrote:

They're claiming "even better than" 5 ns for relative time, which given
the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that claim

to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns

is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a

lot

of antennas.

None of this is getting you to the actual total delay of the antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m

not

sure that

the differential delay is taken out that way. Total delay, not taken

out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X ns,

but it would be part

of any practical system trying to get close to 5 ns absolute accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that

if

the appropriate

one sigma / on a clear day / with the wind in the right direction sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com

wrote:

It's interesting that they talk about the F9 receivers offering 5 ns

absolute time accuracy.  Does anyone know of tests confirming that, and
what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi > On Feb 27, 2021, at 9:41 AM, Dana Whitlow <k8yumdoober@gmail.com> wrote: > > I've long understood that ionospheric delays and variations thereof lead to > *position* > uncertainties in GPS navigation receivers, to the tune of perhaps 10m > (2DRMS IIRC)., > and that these are said to constitute the single largest GPS error source. > > Q1: Would this not imply timing errors of comparable magnitude (10's of > nsec) > for a single band GPS? Once all the signals “hit” the antenna, the delays are mostly common mode. Instead of showing up as a position error, they show up as an error in the time estimate. Since time is one of the things you estimate in the solution (along with X,Y, and Z) it get’s it’s own independent solution. > > Q2: Why have I not seen this issue raised in connection with the present > discussion > about achievable absolute timing accuracy? GPS time transfer is often done to the sub-ns level. There are a number of papers on this. > > Q3: Do any of the "modern" timing GPS receivers available to civilians do > dual-band > reception in a way that includes estimation of (and correction for) > said delays and > their variations? I know that Garmin, for one, is now selling L1/L5 > handheld GPS > receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication > that these > units make any attempt at doing such corrections. Yes, some receivers do an estimate of ionospheric delay based on the variation of that delay with frequency. This does not help with tropospheric delay or all of the various “common mode” issues we have been talking about. It is also unclear how the “unknown” timing variation between the bands due to the antenna impacts these solutions….. Bob > > Dana > > > > On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq <kb8tq@n1k.org> wrote: > >> Hi >> >> The same 20 or so ns delay in a saw would also apply to the >> saw (or tight filter) in some timing antennas. It also would apply >> to the saw(s) in some modules. Even if the tolerance is “only” >> a couple ns on each of them, you *could* have 3 or more in the >> chain. >> >> Lots of numbers to crunch to get to 5 ns “absolute”. One could go >> grab a GPS simulator and start poking. First step would be to find >> a simulator that is spec’d for a < 5 ns tolerance on the PPS into >> GPS out. I do believe that rules out the eBay marvels that some >> of us have lying around ….. >> >> Simpler answer would be a quick “clock trip” with your car full >> of 5071’s …… hour drive over to NIST and then back home. >> That sounds practical for most of us :) :) >> >> Bob >> >>> On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR <jra@febo.com> wrote: >>> >>> A while ago I tried doing a decidedly non-anechoic measurement with a >> VNA exciter going to a 1500 MHz ground plane and the receiver connected to >> the antenna (with a known delay cable) and I got a similar result, but >> there was enough noise that I didn't think I could nail it down to within >> 10 ns. >>> >>> I've also measured GPS antenna splitters and they tend to have 20-ish ns >> delays, mainly due to the SAW filters. I did surgery on an HP splitter to >> remove the filters so it could be used for L1 and L2 and that dropped the >> delay down to only 1 or 2 ns. >>> >>> So there's definitely lots of stuff to calibrate if you want to get >> accurate time transfer. >>> >>> John >>> ---- >>> >>> On 2/26/21 8:02 PM, Michael Wouters wrote: >>>> Typical L1 antenna delays range from 20 to 70 ns. >>>> I know of only one antenna for which a delay is given by the vendor and >> the >>>> technique used was just to measure the electronic delay ie by injecting >> a >>>> signal into the circuit. To do it properly, you need a setup in a >> microwave >>>> anechoic chamber with transmitting antenna etc. The practical difference >>>> may be small though, 1 or 2 ns ( sample of one antenna!). >>>> Cheers >>>> Michael >>>> On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> >> wrote: >>>>> They're claiming "even better than" 5 ns for relative time, which given >>>>> the 4 ns jitter seems at least sort-of reasonable. But until someone >>>>> shows me otherwise, I'm still thinking that getting better than 25 ns >>>>> absolute accuracy is a pretty good day's work. >>>>> >>>>> John >>>>> ---- >>>>> >>>>> On 2/26/21 5:26 PM, Bob kb8tq wrote: >>>>>> Hi >>>>>> >>>>>> I can’t think of many antennas (multi band or single band) that claim >> to >>>>> know their >>>>>> delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns >> is >>>>> quite good. >>>>>> Same thing with delay ripple, you see specs out to around 15 ns on a >> lot >>>>> of antennas. >>>>>> None of this is getting you to the actual total delay of the antenna. >>>>> It’s a pretty good >>>>>> bet that number is a bit larger than either of these. >>>>>> >>>>>> Some of the ripple probably comes out in the standard modeling. I’m >> not >>>>> sure that >>>>>> the differential delay is taken out that way. Total delay, not taken >> out >>>>> in any obvious >>>>>> fashion ( at least that I can see). If the F9 has a built in antenna >>>>> database, that’s not >>>>>> mentioned in the doc’s. Any benefit from the corrections would have to >>>>> be part of >>>>>> post processing. >>>>>> >>>>>> No, that’s not the same as talking about the F9 it’s self doing X ns, >>>>> but it would be part >>>>>> of any practical system trying to get close to 5 ns absolute accuracy. >>>>>> >>>>>> 5 ns *relative* accuracy between two F9’s? I probably could buy that >> if >>>>> the appropriate >>>>>> one sigma / on a clear day / with the wind in the right direction sort >>>>> of qualifiers are >>>>>> attached. >>>>>> >>>>>> Bob >>>>>> >>>>>>> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> >> wrote: >>>>>>> >>>>>>> It's interesting that they talk about the F9 receivers offering 5 ns >>>>> absolute time accuracy. Does anyone know of tests confirming that, and >>>>> what sort of care was required in the setup to get there? >>>>>>> >>>>>>> John >>>>>>> ---- >>>>>>> >>>>>>> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >>>>>>>> FWIW. No detailed content, and a rather quick read. "Five key trends >>>>> in GPS". >>>>>>>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >>>>>>>> _______________________________________________ >>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>>> To unsubscribe, go to >>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>>> and follow the instructions there. >>>>>>> >>>>>>> _______________________________________________ >>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>> To unsubscribe, go to >>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>> and follow the instructions there. >>>>>> >>>>>> >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>> To unsubscribe, go to >>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>> and follow the instructions there. >>>>>> >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>> To unsubscribe, go to >>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>> and follow the instructions there. >>>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
DW
Dana Whitlow
Sat, Feb 27, 2021 4:18 PM

Thanks, Bob.

It seems to me that, depending on the positions of sats visible to one's GPS
antenna and the spatial distribution of free electron density in the
ionosphere,
the ionospheric contribution to position errors could sometimes largely
cancel.
But that observation may (or may not) reflect strongly on one's ability to
get
accurate absolute time from GPS on "average" days.

During my Arecibo Observatory days we used NIST's TMAS service to keep
our H-maser-based station clock synced with UTC.  Our user community
(mainly VLBI and pulsar timing people) seemed pretty satisfied with +/-
100ns
accuracy, so I tried to do better by keeping things well within +/- 50 ns
during
my reign.  IIRC, NIST was claiming that TMAS could produce results mostly
within about +/- 20 ns.

To be honest I'm baffled by how time transfer much better than that could
be achieved in practice.

Regarding Q3, yes I'm aware that some GPS receivers do the estimation of
ionospheric delay.  What I was asking was:  Do any of the relatively
inexpensive
receivers to which we time-nuts have access do this?  Here I'm speaking of
those being sold for no more than a few hundred USD.

Dana

On Sat, Feb 27, 2021 at 9:08 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

On Feb 27, 2021, at 9:41 AM, Dana Whitlow k8yumdoober@gmail.com wrote:

I've long understood that ionospheric delays and variations thereof lead

to

position
uncertainties in GPS navigation receivers, to the tune of perhaps 10m
(2DRMS IIRC).,
and that these are said to constitute the single largest GPS error

source.

Q1: Would this not imply timing errors of comparable magnitude (10's of
nsec)
for a single band GPS?

Once all the signals “hit” the antenna, the delays are mostly common mode.
Instead of showing up as a position error, they show up as an error in the
time estimate. Since time is one of the things you estimate in the
solution
(along with X,Y, and Z) it get’s it’s own independent solution.

Q2: Why have I not seen this issue raised in connection with the present
discussion
about achievable absolute timing accuracy?

GPS time transfer is often done to the sub-ns level. There are a number of
papers on this.

Q3: Do any of the "modern" timing GPS receivers available to civilians do
dual-band
reception in a way that includes estimation of (and correction for)
said delays and
their variations?  I know that Garmin, for one, is now selling L1/L5
handheld GPS
receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication
that these
units make any attempt at doing such corrections.

Yes, some receivers do an estimate of ionospheric delay based on the
variation of that delay with frequency. This does not help with
tropospheric
delay or all of the various “common mode” issues we have been talking
about.
It is also unclear how the “unknown” timing variation between the bands
due to the antenna impacts these solutions…..

Bob

Dana

On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The same 20 or so ns delay in a saw would also apply to the
saw (or tight filter) in some timing antennas. It also would apply
to the saw(s) in some modules. Even if the tolerance is “only”
a couple ns on each of them, you could have 3 or more in the
chain.

Lots of numbers to crunch to get to 5 ns “absolute”. One could go
grab a GPS simulator and start poking. First step would be to find
a simulator that is spec’d for a < 5 ns tolerance on the PPS into
GPS out. I do believe that rules out the eBay marvels that some
of us have lying around …..

Simpler answer would be a quick “clock trip” with your car full
of 5071’s …… hour drive over to NIST and then back home.
That sounds practical for most of us :) :)

Bob

On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR jra@febo.com wrote:

A while ago I tried doing a decidedly non-anechoic measurement with a

VNA exciter going to a 1500 MHz ground plane and the receiver connected

to

the antenna (with a known delay cable) and I got a similar result, but
there was enough noise that I didn't think I could nail it down to

within

10 ns.

I've also measured GPS antenna splitters and they tend to have 20-ish

ns

delays, mainly due to the SAW filters.  I did surgery on an HP splitter

to

remove the filters so it could be used for L1 and L2 and that dropped

the

delay down to only 1 or 2 ns.

So there's definitely lots of stuff to calibrate if you want to get

accurate time transfer.

John

On 2/26/21 8:02 PM, Michael Wouters wrote:

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor

and

the

technique used was just to measure the electronic delay ie by

injecting

a

signal into the circuit. To do it properly, you need a setup in a

microwave

anechoic chamber with transmitting antenna etc. The practical

difference

may be small though, 1 or 2 ns ( sample of one antenna!).
Cheers
Michael
On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com

wrote:

They're claiming "even better than" 5 ns for relative time, which

given

the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that

claim

to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns

is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a

lot

of antennas.

None of this is getting you to the actual total delay of the

antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m

not

sure that

the differential delay is taken out that way. Total delay, not taken

out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have

to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X

ns,

but it would be part

of any practical system trying to get close to 5 ns absolute

accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that

if

the appropriate

one sigma / on a clear day / with the wind in the right direction

sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com

wrote:

It's interesting that they talk about the F9 receivers offering 5

ns

absolute time accuracy.  Does anyone know of tests confirming that,

and

what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key

trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Thanks, Bob. It seems to me that, depending on the positions of sats visible to one's GPS antenna and the spatial distribution of free electron density in the ionosphere, the ionospheric contribution to position errors could sometimes largely cancel. But that observation may (or may not) reflect strongly on one's ability to get accurate absolute time from GPS on "average" days. During my Arecibo Observatory days we used NIST's TMAS service to keep our H-maser-based station clock synced with UTC. Our user community (mainly VLBI and pulsar timing people) seemed pretty satisfied with +/- 100ns accuracy, so I tried to do better by keeping things well within +/- 50 ns during my reign. IIRC, NIST was claiming that TMAS could produce results mostly within about +/- 20 ns. To be honest I'm baffled by how time transfer much better than that could be achieved in practice. Regarding Q3, yes I'm aware that *some* GPS receivers do the estimation of ionospheric delay. What I was asking was: Do any of the relatively inexpensive receivers to which we time-nuts have access do this? Here I'm speaking of those being sold for no more than a few hundred USD. Dana On Sat, Feb 27, 2021 at 9:08 AM Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > > > > On Feb 27, 2021, at 9:41 AM, Dana Whitlow <k8yumdoober@gmail.com> wrote: > > > > I've long understood that ionospheric delays and variations thereof lead > to > > *position* > > uncertainties in GPS navigation receivers, to the tune of perhaps 10m > > (2DRMS IIRC)., > > and that these are said to constitute the single largest GPS error > source. > > > > Q1: Would this not imply timing errors of comparable magnitude (10's of > > nsec) > > for a single band GPS? > > Once all the signals “hit” the antenna, the delays are mostly common mode. > Instead of showing up as a position error, they show up as an error in the > time estimate. Since time is one of the things you estimate in the > solution > (along with X,Y, and Z) it get’s it’s own independent solution. > > > > Q2: Why have I not seen this issue raised in connection with the present > > discussion > > about achievable absolute timing accuracy? > > GPS time transfer is often done to the sub-ns level. There are a number of > papers on this. > > > > > Q3: Do any of the "modern" timing GPS receivers available to civilians do > > dual-band > > reception in a way that includes estimation of (and correction for) > > said delays and > > their variations? I know that Garmin, for one, is now selling L1/L5 > > handheld GPS > > receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication > > that these > > units make any attempt at doing such corrections. > > Yes, some receivers do an estimate of ionospheric delay based on the > variation of that delay with frequency. This does not help with > tropospheric > delay or all of the various “common mode” issues we have been talking > about. > It is also unclear how the “unknown” timing variation between the bands > due to the antenna impacts these solutions….. > > Bob > > > > > Dana > > > > > > > > On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq <kb8tq@n1k.org> wrote: > > > >> Hi > >> > >> The same 20 or so ns delay in a saw would also apply to the > >> saw (or tight filter) in some timing antennas. It also would apply > >> to the saw(s) in some modules. Even if the tolerance is “only” > >> a couple ns on each of them, you *could* have 3 or more in the > >> chain. > >> > >> Lots of numbers to crunch to get to 5 ns “absolute”. One could go > >> grab a GPS simulator and start poking. First step would be to find > >> a simulator that is spec’d for a < 5 ns tolerance on the PPS into > >> GPS out. I do believe that rules out the eBay marvels that some > >> of us have lying around ….. > >> > >> Simpler answer would be a quick “clock trip” with your car full > >> of 5071’s …… hour drive over to NIST and then back home. > >> That sounds practical for most of us :) :) > >> > >> Bob > >> > >>> On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR <jra@febo.com> wrote: > >>> > >>> A while ago I tried doing a decidedly non-anechoic measurement with a > >> VNA exciter going to a 1500 MHz ground plane and the receiver connected > to > >> the antenna (with a known delay cable) and I got a similar result, but > >> there was enough noise that I didn't think I could nail it down to > within > >> 10 ns. > >>> > >>> I've also measured GPS antenna splitters and they tend to have 20-ish > ns > >> delays, mainly due to the SAW filters. I did surgery on an HP splitter > to > >> remove the filters so it could be used for L1 and L2 and that dropped > the > >> delay down to only 1 or 2 ns. > >>> > >>> So there's definitely lots of stuff to calibrate if you want to get > >> accurate time transfer. > >>> > >>> John > >>> ---- > >>> > >>> On 2/26/21 8:02 PM, Michael Wouters wrote: > >>>> Typical L1 antenna delays range from 20 to 70 ns. > >>>> I know of only one antenna for which a delay is given by the vendor > and > >> the > >>>> technique used was just to measure the electronic delay ie by > injecting > >> a > >>>> signal into the circuit. To do it properly, you need a setup in a > >> microwave > >>>> anechoic chamber with transmitting antenna etc. The practical > difference > >>>> may be small though, 1 or 2 ns ( sample of one antenna!). > >>>> Cheers > >>>> Michael > >>>> On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> > >> wrote: > >>>>> They're claiming "even better than" 5 ns for relative time, which > given > >>>>> the 4 ns jitter seems at least sort-of reasonable. But until someone > >>>>> shows me otherwise, I'm still thinking that getting better than 25 ns > >>>>> absolute accuracy is a pretty good day's work. > >>>>> > >>>>> John > >>>>> ---- > >>>>> > >>>>> On 2/26/21 5:26 PM, Bob kb8tq wrote: > >>>>>> Hi > >>>>>> > >>>>>> I can’t think of many antennas (multi band or single band) that > claim > >> to > >>>>> know their > >>>>>> delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns > >> is > >>>>> quite good. > >>>>>> Same thing with delay ripple, you see specs out to around 15 ns on a > >> lot > >>>>> of antennas. > >>>>>> None of this is getting you to the actual total delay of the > antenna. > >>>>> It’s a pretty good > >>>>>> bet that number is a bit larger than either of these. > >>>>>> > >>>>>> Some of the ripple probably comes out in the standard modeling. I’m > >> not > >>>>> sure that > >>>>>> the differential delay is taken out that way. Total delay, not taken > >> out > >>>>> in any obvious > >>>>>> fashion ( at least that I can see). If the F9 has a built in antenna > >>>>> database, that’s not > >>>>>> mentioned in the doc’s. Any benefit from the corrections would have > to > >>>>> be part of > >>>>>> post processing. > >>>>>> > >>>>>> No, that’s not the same as talking about the F9 it’s self doing X > ns, > >>>>> but it would be part > >>>>>> of any practical system trying to get close to 5 ns absolute > accuracy. > >>>>>> > >>>>>> 5 ns *relative* accuracy between two F9’s? I probably could buy that > >> if > >>>>> the appropriate > >>>>>> one sigma / on a clear day / with the wind in the right direction > sort > >>>>> of qualifiers are > >>>>>> attached. > >>>>>> > >>>>>> Bob > >>>>>> > >>>>>>> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> > >> wrote: > >>>>>>> > >>>>>>> It's interesting that they talk about the F9 receivers offering 5 > ns > >>>>> absolute time accuracy. Does anyone know of tests confirming that, > and > >>>>> what sort of care was required in the setup to get there? > >>>>>>> > >>>>>>> John > >>>>>>> ---- > >>>>>>> > >>>>>>> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: > >>>>>>>> FWIW. No detailed content, and a rather quick read. "Five key > trends > >>>>> in GPS". > >>>>>>>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps > >>>>>>>> _______________________________________________ > >>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>>>> To unsubscribe, go to > >>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>>>> and follow the instructions there. > >>>>>>> > >>>>>>> _______________________________________________ > >>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>>> To unsubscribe, go to > >>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>>> and follow the instructions there. > >>>>>> > >>>>>> > >>>>>> _______________________________________________ > >>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>> To unsubscribe, go to > >>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>> and follow the instructions there. > >>>>>> > >>>>> > >>>>> _______________________________________________ > >>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>> To unsubscribe, go to > >>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>> and follow the instructions there. > >>>>> > >>>> _______________________________________________ > >>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>> To unsubscribe, go to > >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>> and follow the instructions there. > >>> > >>> _______________________________________________ > >>> time-nuts mailing list -- time-nuts@lists.febo.com > >>> To unsubscribe, go to > >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>> and follow the instructions there. > >> > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe, go to > >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >
LJ
Lux, Jim
Sat, Feb 27, 2021 9:24 PM

On 2/27/21 8:18 AM, Dana Whitlow wrote:

Thanks, Bob.

It seems to me that, depending on the positions of sats visible to one's GPS
antenna and the spatial distribution of free electron density in the
ionosphere,
the ionospheric contribution to position errors could sometimes largely
cancel.
But that observation may (or may not) reflect strongly on one's ability to
get
accurate absolute time from GPS on "average" days.

During my Arecibo Observatory days we used NIST's TMAS service to keep
our H-maser-based station clock synced with UTC.  Our user community
(mainly VLBI and pulsar timing people) seemed pretty satisfied with +/-
100ns
accuracy, so I tried to do better by keeping things well within +/- 50 ns
during
my reign.  IIRC, NIST was claiming that TMAS could produce results mostly
within about +/- 20 ns.

To be honest I'm baffled by how time transfer much better than that could
be achieved in practice.

Regarding Q3, yes I'm aware that some GPS receivers do the estimation of
ionospheric delay.  What I was asking was:  Do any of the relatively
inexpensive
receivers to which we time-nuts have access do this?  Here I'm speaking of
those being sold for no more than a few hundred USD.

Dana

If it does dual frequency, then it probably compensates for the
ionosphere. The algorithm isn't complex, and really, there's no reason
to do dual or multiple frequency otherwise - You can get plenty of
satellites with a L1 only, so the increased number of observables,
alone, isn't a good reason for dual frequency.

On 2/27/21 8:18 AM, Dana Whitlow wrote: > Thanks, Bob. > > It seems to me that, depending on the positions of sats visible to one's GPS > antenna and the spatial distribution of free electron density in the > ionosphere, > the ionospheric contribution to position errors could sometimes largely > cancel. > But that observation may (or may not) reflect strongly on one's ability to > get > accurate absolute time from GPS on "average" days. > > During my Arecibo Observatory days we used NIST's TMAS service to keep > our H-maser-based station clock synced with UTC. Our user community > (mainly VLBI and pulsar timing people) seemed pretty satisfied with +/- > 100ns > accuracy, so I tried to do better by keeping things well within +/- 50 ns > during > my reign. IIRC, NIST was claiming that TMAS could produce results mostly > within about +/- 20 ns. > > To be honest I'm baffled by how time transfer much better than that could > be achieved in practice. > > Regarding Q3, yes I'm aware that *some* GPS receivers do the estimation of > ionospheric delay. What I was asking was: Do any of the relatively > inexpensive > receivers to which we time-nuts have access do this? Here I'm speaking of > those being sold for no more than a few hundred USD. > > Dana > If it does dual frequency, then it probably compensates for the ionosphere. The algorithm isn't complex, and really, there's no reason to do dual or multiple frequency otherwise - You can get plenty of satellites with a L1 only, so the increased number of observables, alone, isn't a good reason for dual frequency.
MD
Magnus Danielson
Sat, Feb 27, 2021 10:15 PM

Hi,

On 2021-02-27 17:18, Dana Whitlow wrote:

Thanks, Bob.

It seems to me that, depending on the positions of sats visible to one's GPS
antenna and the spatial distribution of free electron density in the
ionosphere,
the ionospheric contribution to position errors could sometimes largely
cancel.
But that observation may (or may not) reflect strongly on one's ability to
get
accurate absolute time from GPS on "average" days.

During my Arecibo Observatory days we used NIST's TMAS service to keep
our H-maser-based station clock synced with UTC.  Our user community
(mainly VLBI and pulsar timing people) seemed pretty satisfied with +/-
100ns
accuracy, so I tried to do better by keeping things well within +/- 50 ns
during
my reign.  IIRC, NIST was claiming that TMAS could produce results mostly
within about +/- 20 ns.

To be honest I'm baffled by how time transfer much better than that could
be achieved in practice.

Regarding Q3, yes I'm aware that some GPS receivers do the estimation of
ionospheric delay.  What I was asking was:  Do any of the relatively
inexpensive
receivers to which we time-nuts have access do this?  Here I'm speaking of
those being sold for no more than a few hundred USD.

The uBlox F9 and Septentrio Mosaic does this. They output observables
such that good realtime and postprocessing can be achieved. By doing
double frequency, you can do trivial ionspheric delay observation and
compensation of that observation, such that both L1 and L2 can be
compensated. This way you can compensate with actual observed ionspheric
delay rather than using the Klobuchar model which on average only
removes half the ionspheric error. Also, that you get both code and
phase observations helps. All contributes significantly to push things
down. At the same time, doing this for as many satellites as a L1 only
receiver do provide more observables and more precision. Antenna and
reflections end up being more of a challenge in addition to calibration,
but it is worth the next level of improvement to push the errors towards
1 ns or sub 1 ns that you find in metrology type of comparisons. With
the F9 and Mosaic, the price of such receivers just dropped significantly.

TMAS is built with older type of receiver, but I suspect they will be
interested in upgrading for the next generation, as things push the
limit for what is possible on reasonable budget.

Speaking of which, I need to do some work on that. :)

Cheers,
Magnus

Hi, On 2021-02-27 17:18, Dana Whitlow wrote: > Thanks, Bob. > > It seems to me that, depending on the positions of sats visible to one's GPS > antenna and the spatial distribution of free electron density in the > ionosphere, > the ionospheric contribution to position errors could sometimes largely > cancel. > But that observation may (or may not) reflect strongly on one's ability to > get > accurate absolute time from GPS on "average" days. > > During my Arecibo Observatory days we used NIST's TMAS service to keep > our H-maser-based station clock synced with UTC. Our user community > (mainly VLBI and pulsar timing people) seemed pretty satisfied with +/- > 100ns > accuracy, so I tried to do better by keeping things well within +/- 50 ns > during > my reign. IIRC, NIST was claiming that TMAS could produce results mostly > within about +/- 20 ns. > > To be honest I'm baffled by how time transfer much better than that could > be achieved in practice. > > Regarding Q3, yes I'm aware that *some* GPS receivers do the estimation of > ionospheric delay. What I was asking was: Do any of the relatively > inexpensive > receivers to which we time-nuts have access do this? Here I'm speaking of > those being sold for no more than a few hundred USD. The uBlox F9 and Septentrio Mosaic does this. They output observables such that good realtime and postprocessing can be achieved. By doing double frequency, you can do trivial ionspheric delay observation and compensation of that observation, such that both L1 and L2 can be compensated. This way you can compensate with actual observed ionspheric delay rather than using the Klobuchar model which on average only removes half the ionspheric error. Also, that you get both code and phase observations helps. All contributes significantly to push things down. At the same time, doing this for as many satellites as a L1 only receiver do provide more observables and more precision. Antenna and reflections end up being more of a challenge in addition to calibration, but it is worth the next level of improvement to push the errors towards 1 ns or sub 1 ns that you find in metrology type of comparisons. With the F9 and Mosaic, the price of such receivers just dropped significantly. TMAS is built with older type of receiver, but I suspect they will be interested in upgrading for the next generation, as things push the limit for what is possible on reasonable budget. Speaking of which, I need to do some work on that. :) Cheers, Magnus
TV
Tom Van Baak
Sat, Feb 27, 2021 10:15 PM

Dana,

During my Arecibo Observatory days we used NIST's TMAS service to keep
our H-maser-based station clock synced with UTC.

And before that, the observatory used Tom Clark's Oncore & SHOWTIME and
later Rick Hambly's CNS clock & Tac32Plus, yes?

Rick continues to develop the CNS clock, having switched from the
Motorola and iLotus receivers to u-blox T receivers. His papers are on
cnssys.com or gpstime.com. Check out a recent one like:

"High-accuracy Time and Frequency in VLBI "
https://www.cnssys.com/files/TOW/High-accuracy_Time_and_Frequency_in_VLBI_2019_sem.pdf

Bonus: lots of graphs and photos, masers, receivers, etc.

Main publication page: https://www.cnssys.com/publications.php

Check out the performance he's getting. This is with a 6T. That's not
dual-frequency or multi-constellation. Just plain old L1 GPS. It's way
better than +/- 20 ns. So I'm really confused by what you're saying
below. Did Arecibo get rid of the CNS clocks?

/tvb

On 2/27/2021 8:18 AM, Dana Whitlow wrote:

Thanks, Bob.

It seems to me that, depending on the positions of sats visible to one's GPS
antenna and the spatial distribution of free electron density in the
ionosphere,
the ionospheric contribution to position errors could sometimes largely
cancel.
But that observation may (or may not) reflect strongly on one's ability to
get
accurate absolute time from GPS on "average" days.

During my Arecibo Observatory days we used NIST's TMAS service to keep
our H-maser-based station clock synced with UTC.  Our user community
(mainly VLBI and pulsar timing people) seemed pretty satisfied with +/-
100ns
accuracy, so I tried to do better by keeping things well within +/- 50 ns
during
my reign.  IIRC, NIST was claiming that TMAS could produce results mostly
within about +/- 20 ns.

To be honest I'm baffled by how time transfer much better than that could
be achieved in practice.

Regarding Q3, yes I'm aware that some GPS receivers do the estimation of
ionospheric delay.  What I was asking was:  Do any of the relatively
inexpensive
receivers to which we time-nuts have access do this?  Here I'm speaking of
those being sold for no more than a few hundred USD.

Dana

On Sat, Feb 27, 2021 at 9:08 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

On Feb 27, 2021, at 9:41 AM, Dana Whitlow k8yumdoober@gmail.com wrote:

I've long understood that ionospheric delays and variations thereof lead

to

position
uncertainties in GPS navigation receivers, to the tune of perhaps 10m
(2DRMS IIRC).,
and that these are said to constitute the single largest GPS error

source.

Q1: Would this not imply timing errors of comparable magnitude (10's of
nsec)
for a single band GPS?

Once all the signals “hit” the antenna, the delays are mostly common mode.
Instead of showing up as a position error, they show up as an error in the
time estimate. Since time is one of the things you estimate in the
solution
(along with X,Y, and Z) it get’s it’s own independent solution.

Q2: Why have I not seen this issue raised in connection with the present
discussion
about achievable absolute timing accuracy?

GPS time transfer is often done to the sub-ns level. There are a number of
papers on this.

Q3: Do any of the "modern" timing GPS receivers available to civilians do
dual-band
reception in a way that includes estimation of (and correction for)
said delays and
their variations?  I know that Garmin, for one, is now selling L1/L5
handheld GPS
receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication
that these
units make any attempt at doing such corrections.

Yes, some receivers do an estimate of ionospheric delay based on the
variation of that delay with frequency. This does not help with
tropospheric
delay or all of the various “common mode” issues we have been talking
about.
It is also unclear how the “unknown” timing variation between the bands
due to the antenna impacts these solutions…..

Bob

Dana

On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The same 20 or so ns delay in a saw would also apply to the
saw (or tight filter) in some timing antennas. It also would apply
to the saw(s) in some modules. Even if the tolerance is “only”
a couple ns on each of them, you could have 3 or more in the
chain.

Lots of numbers to crunch to get to 5 ns “absolute”. One could go
grab a GPS simulator and start poking. First step would be to find
a simulator that is spec’d for a < 5 ns tolerance on the PPS into
GPS out. I do believe that rules out the eBay marvels that some
of us have lying around …..

Simpler answer would be a quick “clock trip” with your car full
of 5071’s …… hour drive over to NIST and then back home.
That sounds practical for most of us :) :)

Bob

On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR jra@febo.com wrote:

A while ago I tried doing a decidedly non-anechoic measurement with a

VNA exciter going to a 1500 MHz ground plane and the receiver connected

to

the antenna (with a known delay cable) and I got a similar result, but
there was enough noise that I didn't think I could nail it down to

within

10 ns.

I've also measured GPS antenna splitters and they tend to have 20-ish

ns

delays, mainly due to the SAW filters.  I did surgery on an HP splitter

to

remove the filters so it could be used for L1 and L2 and that dropped

the

delay down to only 1 or 2 ns.

So there's definitely lots of stuff to calibrate if you want to get

accurate time transfer.

John

On 2/26/21 8:02 PM, Michael Wouters wrote:

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor

and

the

technique used was just to measure the electronic delay ie by

injecting

a

signal into the circuit. To do it properly, you need a setup in a

microwave

anechoic chamber with transmitting antenna etc. The practical

difference

may be small though, 1 or 2 ns ( sample of one antenna!).
Cheers
Michael
On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com

wrote:

They're claiming "even better than" 5 ns for relative time, which

given

the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that

claim

to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns

is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a

lot

of antennas.

None of this is getting you to the actual total delay of the

antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m

not

sure that

the differential delay is taken out that way. Total delay, not taken

out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have

to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X

ns,

but it would be part

of any practical system trying to get close to 5 ns absolute

accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that

if

the appropriate

one sigma / on a clear day / with the wind in the right direction

sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com

wrote:

It's interesting that they talk about the F9 receivers offering 5

ns

absolute time accuracy.  Does anyone know of tests confirming that,

and

what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key

trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Dana, > During my Arecibo Observatory days we used NIST's TMAS service to keep > our H-maser-based station clock synced with UTC. And before that, the observatory used Tom Clark's Oncore & SHOWTIME and later Rick Hambly's CNS clock & Tac32Plus, yes? Rick continues to develop the CNS clock, having switched from the Motorola and iLotus receivers to u-blox T receivers. His papers are on cnssys.com or gpstime.com. Check out a recent one like: "High-accuracy Time and Frequency in VLBI " https://www.cnssys.com/files/TOW/High-accuracy_Time_and_Frequency_in_VLBI_2019_sem.pdf Bonus: lots of graphs and photos, masers, receivers, etc. Main publication page: https://www.cnssys.com/publications.php Check out the performance he's getting. This is with a 6T. That's not dual-frequency or multi-constellation. Just plain old L1 GPS. It's way better than +/- 20 ns. So I'm really confused by what you're saying below. Did Arecibo get rid of the CNS clocks? /tvb On 2/27/2021 8:18 AM, Dana Whitlow wrote: > Thanks, Bob. > > It seems to me that, depending on the positions of sats visible to one's GPS > antenna and the spatial distribution of free electron density in the > ionosphere, > the ionospheric contribution to position errors could sometimes largely > cancel. > But that observation may (or may not) reflect strongly on one's ability to > get > accurate absolute time from GPS on "average" days. > > During my Arecibo Observatory days we used NIST's TMAS service to keep > our H-maser-based station clock synced with UTC. Our user community > (mainly VLBI and pulsar timing people) seemed pretty satisfied with +/- > 100ns > accuracy, so I tried to do better by keeping things well within +/- 50 ns > during > my reign. IIRC, NIST was claiming that TMAS could produce results mostly > within about +/- 20 ns. > > To be honest I'm baffled by how time transfer much better than that could > be achieved in practice. > > Regarding Q3, yes I'm aware that *some* GPS receivers do the estimation of > ionospheric delay. What I was asking was: Do any of the relatively > inexpensive > receivers to which we time-nuts have access do this? Here I'm speaking of > those being sold for no more than a few hundred USD. > > Dana > > > > On Sat, Feb 27, 2021 at 9:08 AM Bob kb8tq <kb8tq@n1k.org> wrote: > >> Hi >> >> >> >>> On Feb 27, 2021, at 9:41 AM, Dana Whitlow <k8yumdoober@gmail.com> wrote: >>> >>> I've long understood that ionospheric delays and variations thereof lead >> to >>> *position* >>> uncertainties in GPS navigation receivers, to the tune of perhaps 10m >>> (2DRMS IIRC)., >>> and that these are said to constitute the single largest GPS error >> source. >>> Q1: Would this not imply timing errors of comparable magnitude (10's of >>> nsec) >>> for a single band GPS? >> Once all the signals “hit” the antenna, the delays are mostly common mode. >> Instead of showing up as a position error, they show up as an error in the >> time estimate. Since time is one of the things you estimate in the >> solution >> (along with X,Y, and Z) it get’s it’s own independent solution. >>> Q2: Why have I not seen this issue raised in connection with the present >>> discussion >>> about achievable absolute timing accuracy? >> GPS time transfer is often done to the sub-ns level. There are a number of >> papers on this. >> >>> Q3: Do any of the "modern" timing GPS receivers available to civilians do >>> dual-band >>> reception in a way that includes estimation of (and correction for) >>> said delays and >>> their variations? I know that Garmin, for one, is now selling L1/L5 >>> handheld GPS >>> receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication >>> that these >>> units make any attempt at doing such corrections. >> Yes, some receivers do an estimate of ionospheric delay based on the >> variation of that delay with frequency. This does not help with >> tropospheric >> delay or all of the various “common mode” issues we have been talking >> about. >> It is also unclear how the “unknown” timing variation between the bands >> due to the antenna impacts these solutions….. >> >> Bob >> >>> Dana >>> >>> >>> >>> On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq <kb8tq@n1k.org> wrote: >>> >>>> Hi >>>> >>>> The same 20 or so ns delay in a saw would also apply to the >>>> saw (or tight filter) in some timing antennas. It also would apply >>>> to the saw(s) in some modules. Even if the tolerance is “only” >>>> a couple ns on each of them, you *could* have 3 or more in the >>>> chain. >>>> >>>> Lots of numbers to crunch to get to 5 ns “absolute”. One could go >>>> grab a GPS simulator and start poking. First step would be to find >>>> a simulator that is spec’d for a < 5 ns tolerance on the PPS into >>>> GPS out. I do believe that rules out the eBay marvels that some >>>> of us have lying around ….. >>>> >>>> Simpler answer would be a quick “clock trip” with your car full >>>> of 5071’s …… hour drive over to NIST and then back home. >>>> That sounds practical for most of us :) :) >>>> >>>> Bob >>>> >>>>> On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR <jra@febo.com> wrote: >>>>> >>>>> A while ago I tried doing a decidedly non-anechoic measurement with a >>>> VNA exciter going to a 1500 MHz ground plane and the receiver connected >> to >>>> the antenna (with a known delay cable) and I got a similar result, but >>>> there was enough noise that I didn't think I could nail it down to >> within >>>> 10 ns. >>>>> I've also measured GPS antenna splitters and they tend to have 20-ish >> ns >>>> delays, mainly due to the SAW filters. I did surgery on an HP splitter >> to >>>> remove the filters so it could be used for L1 and L2 and that dropped >> the >>>> delay down to only 1 or 2 ns. >>>>> So there's definitely lots of stuff to calibrate if you want to get >>>> accurate time transfer. >>>>> John >>>>> ---- >>>>> >>>>> On 2/26/21 8:02 PM, Michael Wouters wrote: >>>>>> Typical L1 antenna delays range from 20 to 70 ns. >>>>>> I know of only one antenna for which a delay is given by the vendor >> and >>>> the >>>>>> technique used was just to measure the electronic delay ie by >> injecting >>>> a >>>>>> signal into the circuit. To do it properly, you need a setup in a >>>> microwave >>>>>> anechoic chamber with transmitting antenna etc. The practical >> difference >>>>>> may be small though, 1 or 2 ns ( sample of one antenna!). >>>>>> Cheers >>>>>> Michael >>>>>> On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> >>>> wrote: >>>>>>> They're claiming "even better than" 5 ns for relative time, which >> given >>>>>>> the 4 ns jitter seems at least sort-of reasonable. But until someone >>>>>>> shows me otherwise, I'm still thinking that getting better than 25 ns >>>>>>> absolute accuracy is a pretty good day's work. >>>>>>> >>>>>>> John >>>>>>> ---- >>>>>>> >>>>>>> On 2/26/21 5:26 PM, Bob kb8tq wrote: >>>>>>>> Hi >>>>>>>> >>>>>>>> I can’t think of many antennas (multi band or single band) that >> claim >>>> to >>>>>>> know their >>>>>>>> delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns >>>> is >>>>>>> quite good. >>>>>>>> Same thing with delay ripple, you see specs out to around 15 ns on a >>>> lot >>>>>>> of antennas. >>>>>>>> None of this is getting you to the actual total delay of the >> antenna. >>>>>>> It’s a pretty good >>>>>>>> bet that number is a bit larger than either of these. >>>>>>>> >>>>>>>> Some of the ripple probably comes out in the standard modeling. I’m >>>> not >>>>>>> sure that >>>>>>>> the differential delay is taken out that way. Total delay, not taken >>>> out >>>>>>> in any obvious >>>>>>>> fashion ( at least that I can see). If the F9 has a built in antenna >>>>>>> database, that’s not >>>>>>>> mentioned in the doc’s. Any benefit from the corrections would have >> to >>>>>>> be part of >>>>>>>> post processing. >>>>>>>> >>>>>>>> No, that’s not the same as talking about the F9 it’s self doing X >> ns, >>>>>>> but it would be part >>>>>>>> of any practical system trying to get close to 5 ns absolute >> accuracy. >>>>>>>> 5 ns *relative* accuracy between two F9’s? I probably could buy that >>>> if >>>>>>> the appropriate >>>>>>>> one sigma / on a clear day / with the wind in the right direction >> sort >>>>>>> of qualifiers are >>>>>>>> attached. >>>>>>>> >>>>>>>> Bob >>>>>>>> >>>>>>>>> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> >>>> wrote: >>>>>>>>> It's interesting that they talk about the F9 receivers offering 5 >> ns >>>>>>> absolute time accuracy. Does anyone know of tests confirming that, >> and >>>>>>> what sort of care was required in the setup to get there? >>>>>>>>> John >>>>>>>>> ---- >>>>>>>>> >>>>>>>>> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >>>>>>>>>> FWIW. No detailed content, and a rather quick read. "Five key >> trends >>>>>>> in GPS". >>>>>>>>>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >>>>>>>>>> _______________________________________________ >>>>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>>>>> To unsubscribe, go to >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>>>>> and follow the instructions there. >>>>>>>>> _______________________________________________ >>>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>>>> To unsubscribe, go to >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>>>> and follow the instructions there. >>>>>>>> >>>>>>>> _______________________________________________ >>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>>> To unsubscribe, go to >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>>> and follow the instructions there. >>>>>>>> >>>>>>> _______________________________________________ >>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>> To unsubscribe, go to >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>> and follow the instructions there. >>>>>>> >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>> To unsubscribe, go to >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>> and follow the instructions there. >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>> To unsubscribe, go to >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>> To unsubscribe, go to >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>> and follow the instructions there. >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
BK
Bob kb8tq
Sat, Feb 27, 2021 11:39 PM

Hi

On Feb 27, 2021, at 11:18 AM, Dana Whitlow k8yumdoober@gmail.com wrote:

Thanks, Bob.

It seems to me that, depending on the positions of sats visible to one's GPS
antenna and the spatial distribution of free electron density in the
ionosphere,
the ionospheric contribution to position errors could sometimes largely
cancel.
But that observation may (or may not) reflect strongly on one's ability to
get
accurate absolute time from GPS on "average" days.

During my Arecibo Observatory days we used NIST's TMAS service to keep
our H-maser-based station clock synced with UTC.  Our user community
(mainly VLBI and pulsar timing people) seemed pretty satisfied with +/-
100ns
accuracy, so I tried to do better by keeping things well within +/- 50 ns
during
my reign.  IIRC, NIST was claiming that TMAS could produce results mostly
within about +/- 20 ns.

To be honest I'm baffled by how time transfer much better than that could
be achieved in practice.

One way (mentioned about a month back on the list) is a two way transfer
via satellite. The “delay is equal in both directions” assumption is pretty good
in this case. Once you have that as a baseline, you can measure the performance
of other approaches.

One of many starting points to rumble down this rabbit hole:

https://www.nist.gov/pml/time-and-frequency-division/time-services/common-view-gps-time-transfer https://www.nist.gov/pml/time-and-frequency-division/time-services/common-view-gps-time-transfer

I would suggest starting with David Allan’s paper (referenced in the link above) as a
pretty good starting point.

Bob

Regarding Q3, yes I'm aware that some GPS receivers do the estimation of
ionospheric delay.  What I was asking was:  Do any of the relatively
inexpensive
receivers to which we time-nuts have access do this?  Here I'm speaking of
those being sold for no more than a few hundred USD.

Dana

On Sat, Feb 27, 2021 at 9:08 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

On Feb 27, 2021, at 9:41 AM, Dana Whitlow k8yumdoober@gmail.com wrote:

I've long understood that ionospheric delays and variations thereof lead

to

position
uncertainties in GPS navigation receivers, to the tune of perhaps 10m
(2DRMS IIRC).,
and that these are said to constitute the single largest GPS error

source.

Q1: Would this not imply timing errors of comparable magnitude (10's of
nsec)
for a single band GPS?

Once all the signals “hit” the antenna, the delays are mostly common mode.
Instead of showing up as a position error, they show up as an error in the
time estimate. Since time is one of the things you estimate in the
solution
(along with X,Y, and Z) it get’s it’s own independent solution.

Q2: Why have I not seen this issue raised in connection with the present
discussion
about achievable absolute timing accuracy?

GPS time transfer is often done to the sub-ns level. There are a number of
papers on this.

Q3: Do any of the "modern" timing GPS receivers available to civilians do
dual-band
reception in a way that includes estimation of (and correction for)
said delays and
their variations?  I know that Garmin, for one, is now selling L1/L5
handheld GPS
receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication
that these
units make any attempt at doing such corrections.

Yes, some receivers do an estimate of ionospheric delay based on the
variation of that delay with frequency. This does not help with
tropospheric
delay or all of the various “common mode” issues we have been talking
about.
It is also unclear how the “unknown” timing variation between the bands
due to the antenna impacts these solutions…..

Bob

Dana

On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The same 20 or so ns delay in a saw would also apply to the
saw (or tight filter) in some timing antennas. It also would apply
to the saw(s) in some modules. Even if the tolerance is “only”
a couple ns on each of them, you could have 3 or more in the
chain.

Lots of numbers to crunch to get to 5 ns “absolute”. One could go
grab a GPS simulator and start poking. First step would be to find
a simulator that is spec’d for a < 5 ns tolerance on the PPS into
GPS out. I do believe that rules out the eBay marvels that some
of us have lying around …..

Simpler answer would be a quick “clock trip” with your car full
of 5071’s …… hour drive over to NIST and then back home.
That sounds practical for most of us :) :)

Bob

On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR jra@febo.com wrote:

A while ago I tried doing a decidedly non-anechoic measurement with a

VNA exciter going to a 1500 MHz ground plane and the receiver connected

to

the antenna (with a known delay cable) and I got a similar result, but
there was enough noise that I didn't think I could nail it down to

within

10 ns.

I've also measured GPS antenna splitters and they tend to have 20-ish

ns

delays, mainly due to the SAW filters.  I did surgery on an HP splitter

to

remove the filters so it could be used for L1 and L2 and that dropped

the

delay down to only 1 or 2 ns.

So there's definitely lots of stuff to calibrate if you want to get

accurate time transfer.

John

On 2/26/21 8:02 PM, Michael Wouters wrote:

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor

and

the

technique used was just to measure the electronic delay ie by

injecting

a

signal into the circuit. To do it properly, you need a setup in a

microwave

anechoic chamber with transmitting antenna etc. The practical

difference

may be small though, 1 or 2 ns ( sample of one antenna!).
Cheers
Michael
On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com

wrote:

They're claiming "even better than" 5 ns for relative time, which

given

the 4 ns jitter seems at least sort-of reasonable.  But until someone
shows me otherwise, I'm still thinking that getting better than 25 ns
absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that

claim

to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5 ns

is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns on a

lot

of antennas.

None of this is getting you to the actual total delay of the

antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling. I’m

not

sure that

the differential delay is taken out that way. Total delay, not taken

out

in any obvious

fashion ( at least that I can see). If the F9 has a built in antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would have

to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X

ns,

but it would be part

of any practical system trying to get close to 5 ns absolute

accuracy.

5 ns relative accuracy between two F9’s? I probably could buy that

if

the appropriate

one sigma / on a clear day / with the wind in the right direction

sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com

wrote:

It's interesting that they talk about the F9 receivers offering 5

ns

absolute time accuracy.  Does anyone know of tests confirming that,

and

what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key

trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Hi > On Feb 27, 2021, at 11:18 AM, Dana Whitlow <k8yumdoober@gmail.com> wrote: > > Thanks, Bob. > > It seems to me that, depending on the positions of sats visible to one's GPS > antenna and the spatial distribution of free electron density in the > ionosphere, > the ionospheric contribution to position errors could sometimes largely > cancel. > But that observation may (or may not) reflect strongly on one's ability to > get > accurate absolute time from GPS on "average" days. > > During my Arecibo Observatory days we used NIST's TMAS service to keep > our H-maser-based station clock synced with UTC. Our user community > (mainly VLBI and pulsar timing people) seemed pretty satisfied with +/- > 100ns > accuracy, so I tried to do better by keeping things well within +/- 50 ns > during > my reign. IIRC, NIST was claiming that TMAS could produce results mostly > within about +/- 20 ns. > > To be honest I'm baffled by how time transfer much better than that could > be achieved in practice. One way (mentioned about a month back on the list) is a two way transfer via satellite. The “delay is equal in both directions” assumption is pretty good in this case. Once you have that as a baseline, you can measure the performance of other approaches. One of *many* starting points to rumble down this rabbit hole: https://www.nist.gov/pml/time-and-frequency-division/time-services/common-view-gps-time-transfer <https://www.nist.gov/pml/time-and-frequency-division/time-services/common-view-gps-time-transfer> I would suggest starting with David Allan’s paper (referenced in the link above) as a pretty good starting point. Bob > > Regarding Q3, yes I'm aware that *some* GPS receivers do the estimation of > ionospheric delay. What I was asking was: Do any of the relatively > inexpensive > receivers to which we time-nuts have access do this? Here I'm speaking of > those being sold for no more than a few hundred USD. > > Dana > > > > On Sat, Feb 27, 2021 at 9:08 AM Bob kb8tq <kb8tq@n1k.org> wrote: > >> Hi >> >> >> >>> On Feb 27, 2021, at 9:41 AM, Dana Whitlow <k8yumdoober@gmail.com> wrote: >>> >>> I've long understood that ionospheric delays and variations thereof lead >> to >>> *position* >>> uncertainties in GPS navigation receivers, to the tune of perhaps 10m >>> (2DRMS IIRC)., >>> and that these are said to constitute the single largest GPS error >> source. >>> >>> Q1: Would this not imply timing errors of comparable magnitude (10's of >>> nsec) >>> for a single band GPS? >> >> Once all the signals “hit” the antenna, the delays are mostly common mode. >> Instead of showing up as a position error, they show up as an error in the >> time estimate. Since time is one of the things you estimate in the >> solution >> (along with X,Y, and Z) it get’s it’s own independent solution. >>> >>> Q2: Why have I not seen this issue raised in connection with the present >>> discussion >>> about achievable absolute timing accuracy? >> >> GPS time transfer is often done to the sub-ns level. There are a number of >> papers on this. >> >>> >>> Q3: Do any of the "modern" timing GPS receivers available to civilians do >>> dual-band >>> reception in a way that includes estimation of (and correction for) >>> said delays and >>> their variations? I know that Garmin, for one, is now selling L1/L5 >>> handheld GPS >>> receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication >>> that these >>> units make any attempt at doing such corrections. >> >> Yes, some receivers do an estimate of ionospheric delay based on the >> variation of that delay with frequency. This does not help with >> tropospheric >> delay or all of the various “common mode” issues we have been talking >> about. >> It is also unclear how the “unknown” timing variation between the bands >> due to the antenna impacts these solutions….. >> >> Bob >> >>> >>> Dana >>> >>> >>> >>> On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq <kb8tq@n1k.org> wrote: >>> >>>> Hi >>>> >>>> The same 20 or so ns delay in a saw would also apply to the >>>> saw (or tight filter) in some timing antennas. It also would apply >>>> to the saw(s) in some modules. Even if the tolerance is “only” >>>> a couple ns on each of them, you *could* have 3 or more in the >>>> chain. >>>> >>>> Lots of numbers to crunch to get to 5 ns “absolute”. One could go >>>> grab a GPS simulator and start poking. First step would be to find >>>> a simulator that is spec’d for a < 5 ns tolerance on the PPS into >>>> GPS out. I do believe that rules out the eBay marvels that some >>>> of us have lying around ….. >>>> >>>> Simpler answer would be a quick “clock trip” with your car full >>>> of 5071’s …… hour drive over to NIST and then back home. >>>> That sounds practical for most of us :) :) >>>> >>>> Bob >>>> >>>>> On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR <jra@febo.com> wrote: >>>>> >>>>> A while ago I tried doing a decidedly non-anechoic measurement with a >>>> VNA exciter going to a 1500 MHz ground plane and the receiver connected >> to >>>> the antenna (with a known delay cable) and I got a similar result, but >>>> there was enough noise that I didn't think I could nail it down to >> within >>>> 10 ns. >>>>> >>>>> I've also measured GPS antenna splitters and they tend to have 20-ish >> ns >>>> delays, mainly due to the SAW filters. I did surgery on an HP splitter >> to >>>> remove the filters so it could be used for L1 and L2 and that dropped >> the >>>> delay down to only 1 or 2 ns. >>>>> >>>>> So there's definitely lots of stuff to calibrate if you want to get >>>> accurate time transfer. >>>>> >>>>> John >>>>> ---- >>>>> >>>>> On 2/26/21 8:02 PM, Michael Wouters wrote: >>>>>> Typical L1 antenna delays range from 20 to 70 ns. >>>>>> I know of only one antenna for which a delay is given by the vendor >> and >>>> the >>>>>> technique used was just to measure the electronic delay ie by >> injecting >>>> a >>>>>> signal into the circuit. To do it properly, you need a setup in a >>>> microwave >>>>>> anechoic chamber with transmitting antenna etc. The practical >> difference >>>>>> may be small though, 1 or 2 ns ( sample of one antenna!). >>>>>> Cheers >>>>>> Michael >>>>>> On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> >>>> wrote: >>>>>>> They're claiming "even better than" 5 ns for relative time, which >> given >>>>>>> the 4 ns jitter seems at least sort-of reasonable. But until someone >>>>>>> shows me otherwise, I'm still thinking that getting better than 25 ns >>>>>>> absolute accuracy is a pretty good day's work. >>>>>>> >>>>>>> John >>>>>>> ---- >>>>>>> >>>>>>> On 2/26/21 5:26 PM, Bob kb8tq wrote: >>>>>>>> Hi >>>>>>>> >>>>>>>> I can’t think of many antennas (multi band or single band) that >> claim >>>> to >>>>>>> know their >>>>>>>> delay to < 5 ns. Simply having a *differential* delay spec of < 5 ns >>>> is >>>>>>> quite good. >>>>>>>> Same thing with delay ripple, you see specs out to around 15 ns on a >>>> lot >>>>>>> of antennas. >>>>>>>> None of this is getting you to the actual total delay of the >> antenna. >>>>>>> It’s a pretty good >>>>>>>> bet that number is a bit larger than either of these. >>>>>>>> >>>>>>>> Some of the ripple probably comes out in the standard modeling. I’m >>>> not >>>>>>> sure that >>>>>>>> the differential delay is taken out that way. Total delay, not taken >>>> out >>>>>>> in any obvious >>>>>>>> fashion ( at least that I can see). If the F9 has a built in antenna >>>>>>> database, that’s not >>>>>>>> mentioned in the doc’s. Any benefit from the corrections would have >> to >>>>>>> be part of >>>>>>>> post processing. >>>>>>>> >>>>>>>> No, that’s not the same as talking about the F9 it’s self doing X >> ns, >>>>>>> but it would be part >>>>>>>> of any practical system trying to get close to 5 ns absolute >> accuracy. >>>>>>>> >>>>>>>> 5 ns *relative* accuracy between two F9’s? I probably could buy that >>>> if >>>>>>> the appropriate >>>>>>>> one sigma / on a clear day / with the wind in the right direction >> sort >>>>>>> of qualifiers are >>>>>>>> attached. >>>>>>>> >>>>>>>> Bob >>>>>>>> >>>>>>>>> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> >>>> wrote: >>>>>>>>> >>>>>>>>> It's interesting that they talk about the F9 receivers offering 5 >> ns >>>>>>> absolute time accuracy. Does anyone know of tests confirming that, >> and >>>>>>> what sort of care was required in the setup to get there? >>>>>>>>> >>>>>>>>> John >>>>>>>>> ---- >>>>>>>>> >>>>>>>>> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: >>>>>>>>>> FWIW. No detailed content, and a rather quick read. "Five key >> trends >>>>>>> in GPS". >>>>>>>>>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps >>>>>>>>>> _______________________________________________ >>>>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>>>>> To unsubscribe, go to >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>>>>> and follow the instructions there. >>>>>>>>> >>>>>>>>> _______________________________________________ >>>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>>>> To unsubscribe, go to >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>>>> and follow the instructions there. >>>>>>>> >>>>>>>> >>>>>>>> _______________________________________________ >>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>>> To unsubscribe, go to >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>>> and follow the instructions there. >>>>>>>> >>>>>>> >>>>>>> _______________________________________________ >>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>>> To unsubscribe, go to >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>>> and follow the instructions there. >>>>>>> >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>>> To unsubscribe, go to >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>>> and follow the instructions there. >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>>> To unsubscribe, go to >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>>> and follow the instructions there. >>>> >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@lists.febo.com >>>> To unsubscribe, go to >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>>> and follow the instructions there. >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@lists.febo.com >>> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >>> and follow the instructions there. >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com >> To unsubscribe, go to >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there.
DW
Dana Whitlow
Sun, Feb 28, 2021 1:07 AM

Thanks for the link, Bob.  I got to know both Victor Zhang and Mike
Lombardi during my
stay at Arecibo, but to my regret have never met either in person.

Dana

On Sat, Feb 27, 2021 at 6:01 PM Bob kb8tq kb8tq@n1k.org wrote:

Hi

On Feb 27, 2021, at 11:18 AM, Dana Whitlow k8yumdoober@gmail.com

wrote:

Thanks, Bob.

It seems to me that, depending on the positions of sats visible to one's

GPS

antenna and the spatial distribution of free electron density in the
ionosphere,
the ionospheric contribution to position errors could sometimes largely
cancel.
But that observation may (or may not) reflect strongly on one's ability

to

get
accurate absolute time from GPS on "average" days.

During my Arecibo Observatory days we used NIST's TMAS service to keep
our H-maser-based station clock synced with UTC.  Our user community
(mainly VLBI and pulsar timing people) seemed pretty satisfied with +/-
100ns
accuracy, so I tried to do better by keeping things well within +/- 50 ns
during
my reign.  IIRC, NIST was claiming that TMAS could produce results mostly
within about +/- 20 ns.

To be honest I'm baffled by how time transfer much better than that could
be achieved in practice.

One way (mentioned about a month back on the list) is a two way transfer
via satellite. The “delay is equal in both directions” assumption is
pretty good
in this case. Once you have that as a baseline, you can measure the
performance
of other approaches.

One of many starting points to rumble down this rabbit hole:

https://www.nist.gov/pml/time-and-frequency-division/time-services/common-view-gps-time-transfer
<
https://www.nist.gov/pml/time-and-frequency-division/time-services/common-view-gps-time-transfer

I would suggest starting with David Allan’s paper (referenced in the link
above) as a
pretty good starting point.

Bob

Regarding Q3, yes I'm aware that some GPS receivers do the estimation

of

ionospheric delay.  What I was asking was:  Do any of the relatively
inexpensive
receivers to which we time-nuts have access do this?  Here I'm speaking

of

those being sold for no more than a few hundred USD.

Dana

On Sat, Feb 27, 2021 at 9:08 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

On Feb 27, 2021, at 9:41 AM, Dana Whitlow k8yumdoober@gmail.com

wrote:

I've long understood that ionospheric delays and variations thereof

lead

to

position
uncertainties in GPS navigation receivers, to the tune of perhaps 10m
(2DRMS IIRC).,
and that these are said to constitute the single largest GPS error

source.

Q1: Would this not imply timing errors of comparable magnitude (10's of
nsec)
for a single band GPS?

Once all the signals “hit” the antenna, the delays are mostly common

mode.

Instead of showing up as a position error, they show up as an error in

the

time estimate. Since time is one of the things you estimate in the
solution
(along with X,Y, and Z) it get’s it’s own independent solution.

Q2: Why have I not seen this issue raised in connection with the

present

discussion
about achievable absolute timing accuracy?

GPS time transfer is often done to the sub-ns level. There are a number

of

papers on this.

Q3: Do any of the "modern" timing GPS receivers available to civilians

do

dual-band
reception in a way that includes estimation of (and correction for)
said delays and
their variations?  I know that Garmin, for one, is now selling

L1/L5

handheld GPS
receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication
that these
units make any attempt at doing such corrections.

Yes, some receivers do an estimate of ionospheric delay based on the
variation of that delay with frequency. This does not help with
tropospheric
delay or all of the various “common mode” issues we have been talking
about.
It is also unclear how the “unknown” timing variation between the bands
due to the antenna impacts these solutions…..

Bob

Dana

On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq kb8tq@n1k.org wrote:

Hi

The same 20 or so ns delay in a saw would also apply to the
saw (or tight filter) in some timing antennas. It also would apply
to the saw(s) in some modules. Even if the tolerance is “only”
a couple ns on each of them, you could have 3 or more in the
chain.

Lots of numbers to crunch to get to 5 ns “absolute”. One could go
grab a GPS simulator and start poking. First step would be to find
a simulator that is spec’d for a < 5 ns tolerance on the PPS into
GPS out. I do believe that rules out the eBay marvels that some
of us have lying around …..

Simpler answer would be a quick “clock trip” with your car full
of 5071’s …… hour drive over to NIST and then back home.
That sounds practical for most of us :) :)

Bob

On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR jra@febo.com

wrote:

A while ago I tried doing a decidedly non-anechoic measurement with a

VNA exciter going to a 1500 MHz ground plane and the receiver

connected

to

the antenna (with a known delay cable) and I got a similar result, but
there was enough noise that I didn't think I could nail it down to

within

10 ns.

I've also measured GPS antenna splitters and they tend to have 20-ish

ns

delays, mainly due to the SAW filters.  I did surgery on an HP

splitter

to

remove the filters so it could be used for L1 and L2 and that dropped

the

delay down to only 1 or 2 ns.

So there's definitely lots of stuff to calibrate if you want to get

accurate time transfer.

John

On 2/26/21 8:02 PM, Michael Wouters wrote:

Typical L1 antenna delays range from 20 to 70 ns.
I know of only one antenna for which a delay is given by the vendor

and

the

technique used was just to measure the electronic delay ie by

injecting

a

signal into the circuit. To do it properly, you need a setup in a

microwave

anechoic chamber with transmitting antenna etc. The practical

difference

may be small though, 1 or 2 ns ( sample of one antenna!).
Cheers
Michael
On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR jra@febo.com

wrote:

They're claiming "even better than" 5 ns for relative time, which

given

the 4 ns jitter seems at least sort-of reasonable.  But until

someone

shows me otherwise, I'm still thinking that getting better than 25

ns

absolute accuracy is a pretty good day's work.

John

On 2/26/21 5:26 PM, Bob kb8tq wrote:

Hi

I can’t think of many antennas (multi band or single band) that

claim

to

know their

delay to < 5 ns. Simply having a differential delay spec of < 5

ns

is

quite good.

Same thing with delay ripple, you see specs out to around 15 ns

on a

lot

of antennas.

None of this is getting you to the actual total delay of the

antenna.

It’s a pretty good

bet that number is a bit larger than either of these.

Some of the ripple probably comes out in the standard modeling.

I’m

not

sure that

the differential delay is taken out that way. Total delay, not

taken

out

in any obvious

fashion ( at least that I can see). If the F9 has a built in

antenna

database, that’s not

mentioned in the doc’s. Any benefit from the corrections would

have

to

be part of

post processing.

No, that’s not the same as talking about the F9 it’s self doing X

ns,

but it would be part

of any practical system trying to get close to 5 ns absolute

accuracy.

5 ns relative accuracy between two F9’s? I probably could buy

that

if

the appropriate

one sigma / on a clear day / with the wind in the right direction

sort

of qualifiers are

attached.

Bob

On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR jra@febo.com

wrote:

It's interesting that they talk about the F9 receivers offering 5

ns

absolute time accuracy.  Does anyone know of tests confirming that,

and

what sort of care was required in the setup to get there?

John

On 2/26/21 9:34 AM, Robert LaJeunesse wrote:

FWIW. No detailed content, and a rather quick read. "Five key

trends

in GPS".

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to

and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Thanks for the link, Bob. I got to know both Victor Zhang and Mike Lombardi during my stay at Arecibo, but to my regret have never met either in person. Dana On Sat, Feb 27, 2021 at 6:01 PM Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > > > > On Feb 27, 2021, at 11:18 AM, Dana Whitlow <k8yumdoober@gmail.com> > wrote: > > > > Thanks, Bob. > > > > It seems to me that, depending on the positions of sats visible to one's > GPS > > antenna and the spatial distribution of free electron density in the > > ionosphere, > > the ionospheric contribution to position errors could sometimes largely > > cancel. > > But that observation may (or may not) reflect strongly on one's ability > to > > get > > accurate absolute time from GPS on "average" days. > > > > During my Arecibo Observatory days we used NIST's TMAS service to keep > > our H-maser-based station clock synced with UTC. Our user community > > (mainly VLBI and pulsar timing people) seemed pretty satisfied with +/- > > 100ns > > accuracy, so I tried to do better by keeping things well within +/- 50 ns > > during > > my reign. IIRC, NIST was claiming that TMAS could produce results mostly > > within about +/- 20 ns. > > > > To be honest I'm baffled by how time transfer much better than that could > > be achieved in practice. > > One way (mentioned about a month back on the list) is a two way transfer > via satellite. The “delay is equal in both directions” assumption is > pretty good > in this case. Once you have that as a baseline, you can measure the > performance > of other approaches. > > One of *many* starting points to rumble down this rabbit hole: > > > https://www.nist.gov/pml/time-and-frequency-division/time-services/common-view-gps-time-transfer > < > https://www.nist.gov/pml/time-and-frequency-division/time-services/common-view-gps-time-transfer > > > > I would suggest starting with David Allan’s paper (referenced in the link > above) as a > pretty good starting point. > > Bob > > > > > Regarding Q3, yes I'm aware that *some* GPS receivers do the estimation > of > > ionospheric delay. What I was asking was: Do any of the relatively > > inexpensive > > receivers to which we time-nuts have access do this? Here I'm speaking > of > > those being sold for no more than a few hundred USD. > > > > Dana > > > > > > > > On Sat, Feb 27, 2021 at 9:08 AM Bob kb8tq <kb8tq@n1k.org> wrote: > > > >> Hi > >> > >> > >> > >>> On Feb 27, 2021, at 9:41 AM, Dana Whitlow <k8yumdoober@gmail.com> > wrote: > >>> > >>> I've long understood that ionospheric delays and variations thereof > lead > >> to > >>> *position* > >>> uncertainties in GPS navigation receivers, to the tune of perhaps 10m > >>> (2DRMS IIRC)., > >>> and that these are said to constitute the single largest GPS error > >> source. > >>> > >>> Q1: Would this not imply timing errors of comparable magnitude (10's of > >>> nsec) > >>> for a single band GPS? > >> > >> Once all the signals “hit” the antenna, the delays are mostly common > mode. > >> Instead of showing up as a position error, they show up as an error in > the > >> time estimate. Since time is one of the things you estimate in the > >> solution > >> (along with X,Y, and Z) it get’s it’s own independent solution. > >>> > >>> Q2: Why have I not seen this issue raised in connection with the > present > >>> discussion > >>> about achievable absolute timing accuracy? > >> > >> GPS time transfer is often done to the sub-ns level. There are a number > of > >> papers on this. > >> > >>> > >>> Q3: Do any of the "modern" timing GPS receivers available to civilians > do > >>> dual-band > >>> reception in a way that includes estimation of (and correction for) > >>> said delays and > >>> their variations? I know that Garmin, for one, is now selling > L1/L5 > >>> handheld GPS > >>> receivers (GPSMAP66sr and GPSMAP65s), but I've seen no indication > >>> that these > >>> units make any attempt at doing such corrections. > >> > >> Yes, some receivers do an estimate of ionospheric delay based on the > >> variation of that delay with frequency. This does not help with > >> tropospheric > >> delay or all of the various “common mode” issues we have been talking > >> about. > >> It is also unclear how the “unknown” timing variation between the bands > >> due to the antenna impacts these solutions….. > >> > >> Bob > >> > >>> > >>> Dana > >>> > >>> > >>> > >>> On Sat, Feb 27, 2021 at 7:43 AM Bob kb8tq <kb8tq@n1k.org> wrote: > >>> > >>>> Hi > >>>> > >>>> The same 20 or so ns delay in a saw would also apply to the > >>>> saw (or tight filter) in some timing antennas. It also would apply > >>>> to the saw(s) in some modules. Even if the tolerance is “only” > >>>> a couple ns on each of them, you *could* have 3 or more in the > >>>> chain. > >>>> > >>>> Lots of numbers to crunch to get to 5 ns “absolute”. One could go > >>>> grab a GPS simulator and start poking. First step would be to find > >>>> a simulator that is spec’d for a < 5 ns tolerance on the PPS into > >>>> GPS out. I do believe that rules out the eBay marvels that some > >>>> of us have lying around ….. > >>>> > >>>> Simpler answer would be a quick “clock trip” with your car full > >>>> of 5071’s …… hour drive over to NIST and then back home. > >>>> That sounds practical for most of us :) :) > >>>> > >>>> Bob > >>>> > >>>>> On Feb 26, 2021, at 9:29 PM, John Ackermann N8UR <jra@febo.com> > wrote: > >>>>> > >>>>> A while ago I tried doing a decidedly non-anechoic measurement with a > >>>> VNA exciter going to a 1500 MHz ground plane and the receiver > connected > >> to > >>>> the antenna (with a known delay cable) and I got a similar result, but > >>>> there was enough noise that I didn't think I could nail it down to > >> within > >>>> 10 ns. > >>>>> > >>>>> I've also measured GPS antenna splitters and they tend to have 20-ish > >> ns > >>>> delays, mainly due to the SAW filters. I did surgery on an HP > splitter > >> to > >>>> remove the filters so it could be used for L1 and L2 and that dropped > >> the > >>>> delay down to only 1 or 2 ns. > >>>>> > >>>>> So there's definitely lots of stuff to calibrate if you want to get > >>>> accurate time transfer. > >>>>> > >>>>> John > >>>>> ---- > >>>>> > >>>>> On 2/26/21 8:02 PM, Michael Wouters wrote: > >>>>>> Typical L1 antenna delays range from 20 to 70 ns. > >>>>>> I know of only one antenna for which a delay is given by the vendor > >> and > >>>> the > >>>>>> technique used was just to measure the electronic delay ie by > >> injecting > >>>> a > >>>>>> signal into the circuit. To do it properly, you need a setup in a > >>>> microwave > >>>>>> anechoic chamber with transmitting antenna etc. The practical > >> difference > >>>>>> may be small though, 1 or 2 ns ( sample of one antenna!). > >>>>>> Cheers > >>>>>> Michael > >>>>>> On Sat, 27 Feb 2021 at 11:42 am, John Ackermann N8UR <jra@febo.com> > >>>> wrote: > >>>>>>> They're claiming "even better than" 5 ns for relative time, which > >> given > >>>>>>> the 4 ns jitter seems at least sort-of reasonable. But until > someone > >>>>>>> shows me otherwise, I'm still thinking that getting better than 25 > ns > >>>>>>> absolute accuracy is a pretty good day's work. > >>>>>>> > >>>>>>> John > >>>>>>> ---- > >>>>>>> > >>>>>>> On 2/26/21 5:26 PM, Bob kb8tq wrote: > >>>>>>>> Hi > >>>>>>>> > >>>>>>>> I can’t think of many antennas (multi band or single band) that > >> claim > >>>> to > >>>>>>> know their > >>>>>>>> delay to < 5 ns. Simply having a *differential* delay spec of < 5 > ns > >>>> is > >>>>>>> quite good. > >>>>>>>> Same thing with delay ripple, you see specs out to around 15 ns > on a > >>>> lot > >>>>>>> of antennas. > >>>>>>>> None of this is getting you to the actual total delay of the > >> antenna. > >>>>>>> It’s a pretty good > >>>>>>>> bet that number is a bit larger than either of these. > >>>>>>>> > >>>>>>>> Some of the ripple probably comes out in the standard modeling. > I’m > >>>> not > >>>>>>> sure that > >>>>>>>> the differential delay is taken out that way. Total delay, not > taken > >>>> out > >>>>>>> in any obvious > >>>>>>>> fashion ( at least that I can see). If the F9 has a built in > antenna > >>>>>>> database, that’s not > >>>>>>>> mentioned in the doc’s. Any benefit from the corrections would > have > >> to > >>>>>>> be part of > >>>>>>>> post processing. > >>>>>>>> > >>>>>>>> No, that’s not the same as talking about the F9 it’s self doing X > >> ns, > >>>>>>> but it would be part > >>>>>>>> of any practical system trying to get close to 5 ns absolute > >> accuracy. > >>>>>>>> > >>>>>>>> 5 ns *relative* accuracy between two F9’s? I probably could buy > that > >>>> if > >>>>>>> the appropriate > >>>>>>>> one sigma / on a clear day / with the wind in the right direction > >> sort > >>>>>>> of qualifiers are > >>>>>>>> attached. > >>>>>>>> > >>>>>>>> Bob > >>>>>>>> > >>>>>>>>> On Feb 26, 2021, at 4:27 PM, John Ackermann N8UR <jra@febo.com> > >>>> wrote: > >>>>>>>>> > >>>>>>>>> It's interesting that they talk about the F9 receivers offering 5 > >> ns > >>>>>>> absolute time accuracy. Does anyone know of tests confirming that, > >> and > >>>>>>> what sort of care was required in the setup to get there? > >>>>>>>>> > >>>>>>>>> John > >>>>>>>>> ---- > >>>>>>>>> > >>>>>>>>> On 2/26/21 9:34 AM, Robert LaJeunesse wrote: > >>>>>>>>>> FWIW. No detailed content, and a rather quick read. "Five key > >> trends > >>>>>>> in GPS". > >>>>>>>>>> https://www.u-blox.com/en/blogs/insights/five-key-trends-gps > >>>>>>>>>> _______________________________________________ > >>>>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>>>>>> To unsubscribe, go to > >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>>>>>> and follow the instructions there. > >>>>>>>>> > >>>>>>>>> _______________________________________________ > >>>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>>>>> To unsubscribe, go to > >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>>>>> and follow the instructions there. > >>>>>>>> > >>>>>>>> > >>>>>>>> _______________________________________________ > >>>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>>>> To unsubscribe, go to > >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>>>> and follow the instructions there. > >>>>>>>> > >>>>>>> > >>>>>>> _______________________________________________ > >>>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>>> To unsubscribe, go to > >>>>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>>> and follow the instructions there. > >>>>>>> > >>>>>> _______________________________________________ > >>>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>>> To unsubscribe, go to > >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>>> and follow the instructions there. > >>>>> > >>>>> _______________________________________________ > >>>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>>> To unsubscribe, go to > >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>>> and follow the instructions there. > >>>> > >>>> > >>>> _______________________________________________ > >>>> time-nuts mailing list -- time-nuts@lists.febo.com > >>>> To unsubscribe, go to > >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>>> and follow the instructions there. > >>>> > >>> _______________________________________________ > >>> time-nuts mailing list -- time-nuts@lists.febo.com > >>> To unsubscribe, go to > >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >>> and follow the instructions there. > >> > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@lists.febo.com > >> To unsubscribe, go to > >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@lists.febo.com > > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com > and follow the instructions there. >