volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Re: [volt-nuts] Fluke 720A Kelvin Varley Divider Questions

DF
Dr. Frank Stellmach
Mon, Sep 20, 2010 6:21 PM

Thanks to Greg, Bill, Frank and all the others for your valuable

comments and insight on the Fluke 720A reference divider.

Frank, I will attempt to answer your questions about calibrating the 5440B.
I have the 5440B manual which states, "the recommended calibration procedure
uses the Fluke 732A Direct Voltage Reference Standard and the Fluke 752A Reference Divider
to establish the necessary voltage standards which are then compared to the calibrator with the Fluke 845A Null Detector."

I would really prefer to get my hands on a Fluke 752A because it would provide me the required highly precise 10:1 and 100:1 divider
ratios that would, in combination with the Fluke 732A's 10 V output, provide me highly accurate standard voltages
for all the required cardinal points of 0.1V, 1V, 10V, 100V, and 1,000V.
Unfortunately, the 752A does not seem to be readily available on the used equipment market (e.g., EBay, Craigslist, etc.)
and the Fluke 720A is available.  While the Fluke 720A may be overkill for providing only two precise divider ratios of 10:1 and 100:1,
it appears to be the only option that is readily available.  So, I understand your statement that I "don't need the 720A for calibrating a 5440B",
but I do not understand your statement that I "can't use a 720A".
Maybe I am not understanding a subtle point that you are making and I sincerely would like to understand it if that is the case.
I think I can use the 720A to provide the required precision 10:1 and 100:1 divider ratios specified in the 5440B calibration procedure
since I can't get a 752A.  Am I missing something?

...

Charlie

Charlie,

the 752A or a 3458A are able to calibrate the 5440B correctly, but not the 720A.
What you are missing is the correct interpretation of the 720A specification.

The 720A is specified "0.1ppm linearity error of input" for a setting S of 1.1 ..0.1.
That means, if you apply 10V input voltage ("Input" is the crucial word!), it gives an error of 1µV.
Related to the output (linearity!!!), that means the 0.1x divided output of 1V is unsecure 1µV, i.e. 1ppm of output, or 1ppm of ratio.

That means, that the 1V reference point for calibration of the 5440B is too unprecise, or "on the edge", as the reference should have 1/10
of the precision level of the UUT.

For the 100:1 transfer, 100mV range, the situation is even worse. 4ppm is the precision of the 5440B, plus EMF, but the 720A gives 4,6ppm on the 100:1 ratio only
For the high voltage range  transfers, you also have to add the nonlinearity of the thermal heating in the same manner (power coefficient of linearity .. of input..").

The 1000V =>  10V transfer is precise to 4+2,6ppm = 6,6ppm for the 720A (10W power dissipation!!), for 100V=>  10V it is 1,2ppm (on the edge again).

For high voltage, the 720A resistance is much too low.

The 752A is better suited for that, its divider resistors have 4 MOhm in the 1000V range, giving 250mW only.
This divider is specified for 0.5 and 0.2 ppm ratio (output) accuracy, thermal effects included.
Low voltage transfers can be done even more precise, ca. 0.2 ppm for a 10V =>  100mV, and 0.15ppm for 10V =>  1V transfers.
Those precision levels are at least 5 times better than the specified accuracy of the 5440B ranges - good metrological practice.

For the beginning I recommend two articles where this aspect is described:

Here you can check my statement about the correct calculation of the ratio error of the 720A:

http://assets.fluke.com/appnotes/Calibration/NF_ncsl02.pdf
and
http://assets.fluke.com/download/Calibration/Seminars/TheReferenceMultimeter&RatioMeasurements.pdf
page 11-19.

And this famous one describes the problem of checking the ultra low nonlinearity of the HP3458A, for which the 720A is not precise enough, due to the same root cause.
http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf

See pages 14 and 23 for discussion.

Here it is stated also, that the definitions of nonlinearities are not precisely given, what I can underline from painful experience.

regards Frank.

> Thanks to Greg, Bill, Frank and all the others for your valuable comments and insight on the Fluke 720A reference divider. > Frank, I will attempt to answer your questions about calibrating the 5440B. > I have the 5440B manual which states, "the recommended calibration procedure > uses the Fluke 732A Direct Voltage Reference Standard and the Fluke 752A Reference Divider > to establish the necessary voltage standards which are then compared to the calibrator with the Fluke 845A Null Detector." > I would really prefer to get my hands on a Fluke 752A because it would provide me the required highly precise 10:1 and 100:1 divider > ratios that would, in combination with the Fluke 732A's 10 V output, provide me highly accurate standard voltages > for all the required cardinal points of 0.1V, 1V, 10V, 100V, and 1,000V. > Unfortunately, the 752A does not seem to be readily available on the used equipment market (e.g., EBay, Craigslist, etc.) > and the Fluke 720A is available. While the Fluke 720A may be overkill for providing only two precise divider ratios of 10:1 and 100:1, > it appears to be the only option that is readily available. So, I understand your statement that I "don't need the 720A for calibrating a 5440B", > but I do not understand your statement that I "can't use a 720A". > Maybe I am not understanding a subtle point that you are making and I sincerely would like to understand it if that is the case. > I think I can use the 720A to provide the required precision 10:1 and 100:1 divider ratios specified in the 5440B calibration procedure > since I can't get a 752A. Am I missing something? ... > Charlie Charlie, the 752A or a 3458A are able to calibrate the 5440B correctly, but not the 720A. What you are missing is the correct interpretation of the 720A specification. The 720A is specified "0.1ppm linearity error of input" for a setting S of 1.1 ..0.1. That means, if you apply 10V input voltage ("Input" is the crucial word!), it gives an error of 1µV. Related to the output (linearity!!!), that means the 0.1x divided output of 1V is unsecure 1µV, i.e. 1ppm of output, or 1ppm of ratio. That means, that the 1V reference point for calibration of the 5440B is too unprecise, or "on the edge", as the reference should have 1/10 of the precision level of the UUT. For the 100:1 transfer, 100mV range, the situation is even worse. 4ppm is the precision of the 5440B, plus EMF, but the 720A gives 4,6ppm on the 100:1 ratio only For the high voltage range transfers, you also have to add the nonlinearity of the thermal heating in the same manner (power coefficient of linearity .. of input.."). The 1000V => 10V transfer is precise to 4+2,6ppm = 6,6ppm for the 720A (10W power dissipation!!), for 100V=> 10V it is 1,2ppm (on the edge again). For high voltage, the 720A resistance is much too low. The 752A is better suited for that, its divider resistors have 4 MOhm in the 1000V range, giving 250mW only. This divider is specified for 0.5 and 0.2 ppm ratio (output) accuracy, thermal effects included. Low voltage transfers can be done even more precise, ca. 0.2 ppm for a 10V => 100mV, and 0.15ppm for 10V => 1V transfers. Those precision levels are at least 5 times better than the specified accuracy of the 5440B ranges - good metrological practice. For the beginning I recommend two articles where this aspect is described: Here you can check my statement about the correct calculation of the ratio error of the 720A: http://assets.fluke.com/appnotes/Calibration/NF_ncsl02.pdf and http://assets.fluke.com/download/Calibration/Seminars/TheReferenceMultimeter&RatioMeasurements.pdf page 11-19. And this famous one describes the problem of checking the ultra low nonlinearity of the HP3458A, for which the 720A is not precise enough, due to the same root cause. http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf See pages 14 and 23 for discussion. Here it is stated also, that the definitions of nonlinearities are not precisely given, what I can underline from painful experience. regards Frank.
W
WB6BNQ
Tue, Sep 21, 2010 12:28 AM

Frank,

You need to re-evaluate some of your statements.  Out of 10 volts, 1uv
is 0.1ppm.  I also feel you should re-read, more slowly, the two Fluke
articles you referenced below.

In so doing, keep in mind that Fluke has several objectives in their
presentation.  First is to live up to the ideals of being a top notch
metrology company.  Second, is to sell their newest products.  Third,
is to sell you on why you want to buy their newest products while not
severely detracting from their older products.  In all cases,
specmanship can be your friend.

In the first Fluke article, while trying to demonstrate that the 720 is
worse, they conveniently stated that they did not do reversal on the
720 set up which would have improved the numbers.  Yet they did so on
the 752.  A small slight of hand that skews the analysis for one
reading the article.  Would it have been enough to thwart their intent
?  Well, we cannot be sure as we do not have the extensive LAB
environment that they have.

So on the one hand they try to out sell the 720.  Yet in the second
Fluke article, they tend to contradict themselves by stating that the
720 is better than the 8505A DMM which they are trying to claim can do
the job.  So, which is it ?  Their justification ?  The cost and
complexity of using the 720 and the required knowledge to use it
properly.

It is all specmanship at its finest !  I am not saying their new
equipment is a sham or that it is not worth having.  Just that with
each new run of equipment, the sales department has to find ways of
marketing the product.  Hopefully, the customer is getting what he pays
for in the end.

All of this specmanship race does not address whether or not you truly
need such a level of complexity.  If you are trying to construct a real
(doing business that is, unless you are just filthy rich) top notch
calibration facility, then all of that equipment is needed, including
the 720.  If you are just doing you own HOME arrangement, then a
different set of parameters apply.  By the way, that Josephson Voltage
Standard is a few 100K dollars all by itself.  Lacking such an item
makes it near impossible to prove all those specmanship numbers.

So, what we have not asked Charlie yet is what is his arrangement ?
Does he have a temperature and humidity controlled space worthy of
competing with Fluke ?  OR, is he just fooling around at home ?

If he is just fooling around at home, even if he has SOME control of
his space, then he is kidding himself on the level he is going to
achieve.  Not that he shouldnt try to do better all the time, but there
is a cost involved and going that extra millimeter might really be
silly.

Another thing to keep in mind about all those specmanship numbers is
they only apply if you have the same LAB conditions as the original
reference items were calibrated or verified in.  The minute that
reference instrument leaves that environment, all bets are off.

Going back to Charlies original post, he states that this is for his
HOME LAB.  So his interests are to do the best he can.  He has
indicated that the cost of the 752 is too much, so we know either he is
not filthy rich or he is being very responsible with his means of
living.  Truth be known, he would probably be finding a bush to live
under if the wife caught him going overboard.

So, the end result is the 720, and the 750 for that matter, would
satisfy his needs to a given level few truly need.  Forget the
specmanship as metrology is all about knowing your limitations and
applying them properly.

One last point Frank, you mentioned a 10 to 1 ratio for the standard to
the Unit Under Test.  While that is nice, more often then not that is
not achieved when you start playing around at the levels that Fluke
does, even in their own LAB.  The same holds true for NIST, for in some
cases the ratio is down to 2 to 1 or even less.

Bill....WB6BNQ

"Dr. Frank Stellmach" wrote:

 Charlie,

 the 752A or a 3458A are able to calibrate the 5440B correctly, but
 not the 720A.
 What you are missing is the correct interpretation of the 720A
 specification.

 The 720A is specified "0.1ppm linearity error of input" for a
 setting S of 1.1 ..0.1.
 That means, if you apply 10V input voltage ("Input" is the crucial
 word!), it gives an error of 1µV.
 Related to the output (linearity!!!), that means the 0.1x divided
 output of 1V is unsecure 1µV, i.e. 1ppm of output, or 1ppm of ratio.

 That means, that the 1V reference point for calibration of the 5440B
 is too unprecise, or "on the edge", as the reference should have
 1/10
 of the precision level of the UUT.

 For the 100:1 transfer, 100mV range, the situation is even worse.
 4ppm is the precision of the 5440B, plus EMF, but the 720A gives
 4,6ppm on the 100:1 ratio only
 For the high voltage range  transfers, you also have to add the
 nonlinearity of the thermal heating in the same manner (power
 coefficient of linearity .. of input..").

 The 1000V =>  10V transfer is precise to 4+2,6ppm = 6,6ppm for the
 720A (10W power dissipation!!), for 100V=>  10V it is 1,2ppm (on the
 edge again).

 For high voltage, the 720A resistance is much too low.

 The 752A is better suited for that, its divider resistors have 4
 MOhm in the 1000V range, giving 250mW only.
 This divider is specified for 0.5 and 0.2 ppm ratio (output)
 accuracy, thermal effects included.
 Low voltage transfers can be done even more precise, ca. 0.2 ppm for
 a 10V =>  100mV, and 0.15ppm for 10V =>  1V transfers.
 Those precision levels are at least 5 times better than the
 specified accuracy of the 5440B ranges - good metrological practice.

 For the beginning I recommend two articles where this aspect is
 described:

 Here you can check my statement about the correct calculation of the
 ratio error of the 720A:

 [1]http://assets.fluke.com/appnotes/Calibration/NF_ncsl02.pdf
 and
 [2]http://assets.fluke.com/download/Calibration/Seminars/TheReferenc
 eMultimeter&RatioMeasurements.pdf
 page 11-19.

 And this famous one describes the problem of checking the ultra low
 nonlinearity of the HP3458A, for which the 720A is not precise
 enough, due to the same root cause.
 [3]http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf

 See pages 14 and 23 for discussion.

 Here it is stated also, that the definitions of nonlinearities are
 not precisely given, what I can underline from painful experience.

 regards Frank.

 _______________________________________________
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 [4]https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

References

  1. http://assets.fluke.com/appnotes/Calibration/NF_ncsl02.pdf
  2. http://assets.fluke.com/download/Calibration/Seminars/TheReferenceMultimeter&RatioMeasurements.pdf
  3. http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf
  4. https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
Frank, You need to re-evaluate some of your statements. Out of 10 volts, 1uv is 0.1ppm. I also feel you should re-read, more slowly, the two Fluke articles you referenced below. In so doing, keep in mind that Fluke has several objectives in their presentation. First is to live up to the ideals of being a top notch metrology company. Second, is to sell their newest products. Third, is to sell you on why you want to buy their newest products while not severely detracting from their older products. In all cases, specmanship can be your friend. In the first Fluke article, while trying to demonstrate that the 720 is worse, they conveniently stated that they did not do reversal on the 720 set up which would have improved the numbers. Yet they did so on the 752. A small slight of hand that skews the analysis for one reading the article. Would it have been enough to thwart their intent ? Well, we cannot be sure as we do not have the extensive LAB environment that they have. So on the one hand they try to out sell the 720. Yet in the second Fluke article, they tend to contradict themselves by stating that the 720 is better than the 8505A DMM which they are trying to claim can do the job. So, which is it ? Their justification ? The cost and complexity of using the 720 and the required knowledge to use it properly. It is all specmanship at its finest ! I am not saying their new equipment is a sham or that it is not worth having. Just that with each new run of equipment, the sales department has to find ways of marketing the product. Hopefully, the customer is getting what he pays for in the end. All of this specmanship race does not address whether or not you truly need such a level of complexity. If you are trying to construct a real (doing business that is, unless you are just filthy rich) top notch calibration facility, then all of that equipment is needed, including the 720. If you are just doing you own HOME arrangement, then a different set of parameters apply. By the way, that Josephson Voltage Standard is a few 100K dollars all by itself. Lacking such an item makes it near impossible to prove all those specmanship numbers. So, what we have not asked Charlie yet is what is his arrangement ? Does he have a temperature and humidity controlled space worthy of competing with Fluke ? OR, is he just fooling around at home ? If he is just fooling around at home, even if he has SOME control of his space, then he is kidding himself on the level he is going to achieve. Not that he shouldnt try to do better all the time, but there is a cost involved and going that extra millimeter might really be silly. Another thing to keep in mind about all those specmanship numbers is they only apply if you have the same LAB conditions as the original reference items were calibrated or verified in. The minute that reference instrument leaves that environment, all bets are off. Going back to Charlies original post, he states that this is for his HOME LAB. So his interests are to do the best he can. He has indicated that the cost of the 752 is too much, so we know either he is not filthy rich or he is being very responsible with his means of living. Truth be known, he would probably be finding a bush to live under if the wife caught him going overboard. So, the end result is the 720, and the 750 for that matter, would satisfy his needs to a given level few truly need. Forget the specmanship as metrology is all about knowing your limitations and applying them properly. One last point Frank, you mentioned a 10 to 1 ratio for the standard to the Unit Under Test. While that is nice, more often then not that is not achieved when you start playing around at the levels that Fluke does, even in their own LAB. The same holds true for NIST, for in some cases the ratio is down to 2 to 1 or even less. Bill....WB6BNQ "Dr. Frank Stellmach" wrote: Charlie, the 752A or a 3458A are able to calibrate the 5440B correctly, but not the 720A. What you are missing is the correct interpretation of the 720A specification. The 720A is specified "0.1ppm linearity error of input" for a setting S of 1.1 ..0.1. That means, if you apply 10V input voltage ("Input" is the crucial word!), it gives an error of 1µV. Related to the output (linearity!!!), that means the 0.1x divided output of 1V is unsecure 1µV, i.e. 1ppm of output, or 1ppm of ratio. That means, that the 1V reference point for calibration of the 5440B is too unprecise, or "on the edge", as the reference should have 1/10 of the precision level of the UUT. For the 100:1 transfer, 100mV range, the situation is even worse. 4ppm is the precision of the 5440B, plus EMF, but the 720A gives 4,6ppm on the 100:1 ratio only For the high voltage range transfers, you also have to add the nonlinearity of the thermal heating in the same manner (power coefficient of linearity .. of input.."). The 1000V => 10V transfer is precise to 4+2,6ppm = 6,6ppm for the 720A (10W power dissipation!!), for 100V=> 10V it is 1,2ppm (on the edge again). For high voltage, the 720A resistance is much too low. The 752A is better suited for that, its divider resistors have 4 MOhm in the 1000V range, giving 250mW only. This divider is specified for 0.5 and 0.2 ppm ratio (output) accuracy, thermal effects included. Low voltage transfers can be done even more precise, ca. 0.2 ppm for a 10V => 100mV, and 0.15ppm for 10V => 1V transfers. Those precision levels are at least 5 times better than the specified accuracy of the 5440B ranges - good metrological practice. For the beginning I recommend two articles where this aspect is described: Here you can check my statement about the correct calculation of the ratio error of the 720A: [1]http://assets.fluke.com/appnotes/Calibration/NF_ncsl02.pdf and [2]http://assets.fluke.com/download/Calibration/Seminars/TheReferenc eMultimeter&RatioMeasurements.pdf page 11-19. And this famous one describes the problem of checking the ultra low nonlinearity of the HP3458A, for which the 720A is not precise enough, due to the same root cause. [3]http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf See pages 14 and 23 for discussion. Here it is stated also, that the definitions of nonlinearities are not precisely given, what I can underline from painful experience. regards Frank. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to [4]https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. References 1. http://assets.fluke.com/appnotes/Calibration/NF_ncsl02.pdf 2. http://assets.fluke.com/download/Calibration/Seminars/TheReferenceMultimeter&RatioMeasurements.pdf 3. http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf 4. https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
GB
Greg Burnett
Wed, Sep 22, 2010 4:39 AM

Bill,

Yes, "Out of 10 volts, 1uv is 0.1ppm." But take another look at Frank's statement. He's referring to 0.1x divided output uncertainty, for which the uncertainty becomes 1ppm, as he states. (This is for the case where Input = 10Vdc and Output = 1Vdc.) The key to understanding this is that the 720A's linearity is spec'd relative to Input, not output (as Frank is careful to point out).

Greg

----- Original Message -----
From: "WB6BNQ"
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Monday, September 20, 2010 6:28 PM
Subject: Re: [volt-nuts] Fluke 720A Kelvin Varley Divider Questions

Frank,

...Out of 10 volts, 1uv is 0.1ppm.

------clip-----

Bill....WB6BNQ


Frank wrote:
The 720A is specified "0.1ppm linearity error of input" for a setting S of 1.1 ..0.1.
That means, if you apply 10V input voltage ("Input" is the crucial word!), it gives an error of 1µV.
Related to the output (linearity!!!), that means the 0.1x divided output of 1V is unsecure 1µV, i.e. 1ppm of output, or 1ppm of ratio.

Bill, Yes, "Out of 10 volts, 1uv is 0.1ppm." But take another look at Frank's statement. He's referring to 0.1x divided output uncertainty, for which the uncertainty becomes 1ppm, as he states. (This is for the case where Input = 10Vdc and Output = 1Vdc.) The key to understanding this is that the 720A's linearity is spec'd relative to *Input*, not output (as Frank is careful to point out). Greg ----- Original Message ----- From: "WB6BNQ" To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Monday, September 20, 2010 6:28 PM Subject: Re: [volt-nuts] Fluke 720A Kelvin Varley Divider Questions Frank, ...Out of 10 volts, 1uv is 0.1ppm. ------clip----- Bill....WB6BNQ -------------------------------------- Frank wrote: The 720A is specified "0.1ppm linearity error of input" for a setting S of 1.1 ..0.1. That means, if you apply 10V input voltage ("Input" is the crucial word!), it gives an error of 1µV. Related to the output (linearity!!!), that means the 0.1x divided output of 1V is unsecure 1µV, i.e. 1ppm of output, or 1ppm of ratio.