Re: My AC measurement project & question

HM
Hal Murray
Mon, Jan 24, 2022 5:48 AM

If the transformer works well up to 600Hz, isn���t that high enough to capture
anything that actually is grid related (as opposed to local to your home /
neighborhood )?

What's the frequency response of the transformer
on the pole outside my house
at the nearby substation
at a big substation

I remember that back in the days of tubes where audio amplifiers had a
transformer to drive a speaker, there was a lot of iron in the transformer.  I
think that was for the low frequency.  I don't remember anything about high
frequencies.  Was that because there wasn't a problem or I wasn't paying
attention?

What is the frequency response of a wall wart size transformer?


Is the frequency response of transmission lines interesting?

--
These are my opinions.  I hate spam.

kb8tq@n1k.org said: > If the transformer works well up to 600Hz, isn���t that high enough to capture > anything that actually is grid related (as opposed to local to your home / > neighborhood )? What's the frequency response of the transformer on the pole outside my house at the nearby substation at a big substation I remember that back in the days of tubes where audio amplifiers had a transformer to drive a speaker, there was a lot of iron in the transformer. I think that was for the low frequency. I don't remember anything about high frequencies. Was that because there wasn't a problem or I wasn't paying attention? What is the frequency response of a wall wart size transformer? --------- Is the frequency response of transmission lines interesting? -- These are my opinions. I hate spam.
CH
Chris Hastreiter
Mon, Jan 24, 2022 1:18 PM

I work in the distribution grid protection equipment business, and our
products include measurement capability up to the 16th harmonic for both
the voltage and current measurements made on the line.  For classic grid
protection primarily the fundamental and second harmonic are used, but the
higher harmonics are being used more now for high impedance arc fault
detection (think California wildfires) and understanding the operating
behavior of distributed generation resources.

The frequency response characteristics of all the related transformers
certainly aren't linear, but most of these algorithms are looking for the
presence of higher harmonic content rather than trying to measure specific
magnitude values for them.  So yes, higher frequency content can certainly
be on the grid at least at the distribution level, but in a residential
setting much of the harmonic content present will be caused by the
conducted emissions from local devices through their power supplies.

Chris

On Mon, Jan 24, 2022 at 12:04 AM Hal Murray halmurray@sonic.net wrote:

If the transformer works well up to 600Hz, isn???t that high enough to

capture

anything that actually is grid related (as opposed to local to your home

/

neighborhood )?

What's the frequency response of the transformer
on the pole outside my house
at the nearby substation
at a big substation

I remember that back in the days of tubes where audio amplifiers had a
transformer to drive a speaker, there was a lot of iron in the
transformer.  I
think that was for the low frequency.  I don't remember anything about
high
frequencies.  Was that because there wasn't a problem or I wasn't paying
attention?

What is the frequency response of a wall wart size transformer?


Is the frequency response of transmission lines interesting?

--
These are my opinions.  I hate spam.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send
an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

I work in the distribution grid protection equipment business, and our products include measurement capability up to the 16th harmonic for both the voltage and current measurements made on the line. For classic grid protection primarily the fundamental and second harmonic are used, but the higher harmonics are being used more now for high impedance arc fault detection (think California wildfires) and understanding the operating behavior of distributed generation resources. The frequency response characteristics of all the related transformers certainly aren't linear, but most of these algorithms are looking for the presence of higher harmonic content rather than trying to measure specific magnitude values for them. So yes, higher frequency content can certainly be on the grid at least at the distribution level, but in a residential setting much of the harmonic content present will be caused by the conducted emissions from local devices through their power supplies. Chris On Mon, Jan 24, 2022 at 12:04 AM Hal Murray <halmurray@sonic.net> wrote: > > kb8tq@n1k.org said: > > If the transformer works well up to 600Hz, isn???t that high enough to > capture > > anything that actually is grid related (as opposed to local to your home > / > > neighborhood )? > > What's the frequency response of the transformer > on the pole outside my house > at the nearby substation > at a big substation > > I remember that back in the days of tubes where audio amplifiers had a > transformer to drive a speaker, there was a lot of iron in the > transformer. I > think that was for the low frequency. I don't remember anything about > high > frequencies. Was that because there wasn't a problem or I wasn't paying > attention? > > What is the frequency response of a wall wart size transformer? > > --------- > > Is the frequency response of transmission lines interesting? > > -- > These are my opinions. I hate spam. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send > an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
EK
Erik Kaashoek
Mon, Jan 24, 2022 1:43 PM

For some project I'm trying to establish the short term accuracy of a
frequency counter versus the gate time.
As using the Allan Deviation for this type of measurement did lead to
extensive discussion over the validity of using ADEV for measuring the
short term performance of a counter I tried to find a different, but still
relevant way to establish the performance.
To exclude as much as possible external and long term factors I'm using a
single fairly stable OCXO (short term error below 1e-10) to output 10 MHz.
This 10 MHz goes into an SI5351 as reference for its PLL and the SI5351
outputs two frequencies from the same VCO, one at 10 MHz into input A of
the counter and one at 10.00003319 MHz into input B of the counter. The
counter is setup to measure the ratio of A/B and to display the STDDEV of
the ratio over n=100. The STDEV of counter B is calculated as the square
root of  ( (the sum of the squares of the difference between the measured
ratio and the average ratio ) divided by the number of measurements )
I'm aware the SI5351 uses a fractional divider but I hope the impact is
below the measurement accuracy required.
Doing this test with two counter gave these results:

Counter A
Gate time :  STDDEV
1 s : 1.0-10
0.1 s : 1.0e-9
0.02 s : 6.5e-9

Counter B
Gate time :  STDDEV
1 s : 1.3--9
0.1 s : 1.5e-8
0.02 s : 1.4e-7

The results have been verified by performing multiple measurements. Counter
A and B are both fractional counters that use interpolation.

The manual of the Agilent 53132A specifies the worst case RMS  error of a
frequency measurement for different gate times and an input frequency of
10MHz as:

Agilent 53132A
Gate time : Max RMS error (estimated)
1 s :  2e-10
0.1 s : 2e-9
0.02 s :  5e-8

Assuming the RMS error and the STDDEV are the same the steps with gate time
change of the Agilent and Counter A seem to be comparable but Counter B
behaves a bit different for 0.02 s gate time.

This leads me to the following questions:
Is measuring the STDDEV of the ratio of two input frequencies derived from
the same timebase a valid way to assess the short term measurement accuracy
of a frequency counter?
If not, how should this be done?
If yes, do the numbers I'v listed above make sense?

For some project I'm trying to establish the short term accuracy of a frequency counter versus the gate time. As using the Allan Deviation for this type of measurement did lead to extensive discussion over the validity of using ADEV for measuring the short term performance of a counter I tried to find a different, but still relevant way to establish the performance. To exclude as much as possible external and long term factors I'm using a single fairly stable OCXO (short term error below 1e-10) to output 10 MHz. This 10 MHz goes into an SI5351 as reference for its PLL and the SI5351 outputs two frequencies from the same VCO, one at 10 MHz into input A of the counter and one at 10.00003319 MHz into input B of the counter. The counter is setup to measure the ratio of A/B and to display the STDDEV of the ratio over n=100. The STDEV of counter B is calculated as the square root of ( (the sum of the squares of the difference between the measured ratio and the average ratio ) divided by the number of measurements ) I'm aware the SI5351 uses a fractional divider but I hope the impact is below the measurement accuracy required. Doing this test with two counter gave these results: Counter A Gate time : STDDEV 1 s : 1.0-10 0.1 s : 1.0e-9 0.02 s : 6.5e-9 Counter B Gate time : STDDEV 1 s : 1.3--9 0.1 s : 1.5e-8 0.02 s : 1.4e-7 The results have been verified by performing multiple measurements. Counter A and B are both fractional counters that use interpolation. The manual of the Agilent 53132A specifies the worst case RMS error of a frequency measurement for different gate times and an input frequency of 10MHz as: Agilent 53132A Gate time : Max RMS error (estimated) 1 s : 2e-10 0.1 s : 2e-9 0.02 s : 5e-8 Assuming the RMS error and the STDDEV are the same the steps with gate time change of the Agilent and Counter A seem to be comparable but Counter B behaves a bit different for 0.02 s gate time. This leads me to the following questions: Is measuring the STDDEV of the ratio of two input frequencies derived from the same timebase a valid way to assess the short term measurement accuracy of a frequency counter? If not, how should this be done? If yes, do the numbers I'v listed above make sense?
BK
Bob kb8tq
Mon, Jan 24, 2022 2:47 PM

Hi

First off, yes, standard deviation is a pretty good way to look at what a counter
is doing. Reducing the answer to time ( = picoseconds ) is usually the easy way
to look at the data.

Next up, counters have a lot of things that impact what they do. The slew rate
of the input signal is a big one on most counters ( = square waves and sine
waves will give you different answers).  It is best to do your testing with the
type of signal you are most likely to use.

Some counters do very odd things when the input and reference are tightly coupled.
An SR620 counting it’s reference output is one good example of this. Best practice
is to run two independent sources. One supplies the reference, a second one
generates the test signal.

HP counters (and likely some others) have interesting “dead spots” at 10 MHz, at
10MHz / N and 10 MHz * N. to get the best performance numbers, test with a signal
that is not in one of these regions.

It is worth looking at the ADEV / phase noise of your test signal. You can indeed
get into trouble there…..

Fun !!!

Bob

On Jan 24, 2022, at 8:43 AM, Erik Kaashoek erik@kaashoek.com wrote:

For some project I'm trying to establish the short term accuracy of a
frequency counter versus the gate time.
As using the Allan Deviation for this type of measurement did lead to
extensive discussion over the validity of using ADEV for measuring the
short term performance of a counter I tried to find a different, but still
relevant way to establish the performance.
To exclude as much as possible external and long term factors I'm using a
single fairly stable OCXO (short term error below 1e-10) to output 10 MHz.
This 10 MHz goes into an SI5351 as reference for its PLL and the SI5351
outputs two frequencies from the same VCO, one at 10 MHz into input A of
the counter and one at 10.00003319 MHz into input B of the counter. The
counter is setup to measure the ratio of A/B and to display the STDDEV of
the ratio over n=100. The STDEV of counter B is calculated as the square
root of  ( (the sum of the squares of the difference between the measured
ratio and the average ratio ) divided by the number of measurements )
I'm aware the SI5351 uses a fractional divider but I hope the impact is
below the measurement accuracy required.
Doing this test with two counter gave these results:

Counter A
Gate time :  STDDEV
1 s : 1.0-10
0.1 s : 1.0e-9
0.02 s : 6.5e-9

Counter B
Gate time :  STDDEV
1 s : 1.3--9
0.1 s : 1.5e-8
0.02 s : 1.4e-7

The results have been verified by performing multiple measurements. Counter
A and B are both fractional counters that use interpolation.

The manual of the Agilent 53132A specifies the worst case RMS  error of a
frequency measurement for different gate times and an input frequency of
10MHz as:

Agilent 53132A
Gate time : Max RMS error (estimated)
1 s :  2e-10
0.1 s : 2e-9
0.02 s :  5e-8

Assuming the RMS error and the STDDEV are the same the steps with gate time
change of the Agilent and Counter A seem to be comparable but Counter B
behaves a bit different for 0.02 s gate time.

This leads me to the following questions:
Is measuring the STDDEV of the ratio of two input frequencies derived from
the same timebase a valid way to assess the short term measurement accuracy
of a frequency counter?
If not, how should this be done?
If yes, do the numbers I'v listed above make sense?


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Hi First off, yes, standard deviation is a pretty good way to look at what a counter is doing. Reducing the answer to time ( = picoseconds ) is usually the easy way to look at the data. Next up, counters have a *lot* of things that impact what they do. The slew rate of the input signal is a big one on most counters ( = square waves and sine waves will give you different answers). It is best to do your testing with the type of signal you are most likely to use. Some counters do very odd things when the input and reference are tightly coupled. An SR620 counting it’s reference output is one good example of this. Best practice is to run two independent sources. One supplies the reference, a second one generates the test signal. HP counters (and likely some others) have interesting “dead spots” at 10 MHz, at 10MHz / N and 10 MHz * N. to get the best performance numbers, test with a signal that is not in one of these regions. It is worth looking at the ADEV / phase noise of your test signal. You can indeed get into trouble there….. Fun !!! Bob > On Jan 24, 2022, at 8:43 AM, Erik Kaashoek <erik@kaashoek.com> wrote: > > For some project I'm trying to establish the short term accuracy of a > frequency counter versus the gate time. > As using the Allan Deviation for this type of measurement did lead to > extensive discussion over the validity of using ADEV for measuring the > short term performance of a counter I tried to find a different, but still > relevant way to establish the performance. > To exclude as much as possible external and long term factors I'm using a > single fairly stable OCXO (short term error below 1e-10) to output 10 MHz. > This 10 MHz goes into an SI5351 as reference for its PLL and the SI5351 > outputs two frequencies from the same VCO, one at 10 MHz into input A of > the counter and one at 10.00003319 MHz into input B of the counter. The > counter is setup to measure the ratio of A/B and to display the STDDEV of > the ratio over n=100. The STDEV of counter B is calculated as the square > root of ( (the sum of the squares of the difference between the measured > ratio and the average ratio ) divided by the number of measurements ) > I'm aware the SI5351 uses a fractional divider but I hope the impact is > below the measurement accuracy required. > Doing this test with two counter gave these results: > > Counter A > Gate time : STDDEV > 1 s : 1.0-10 > 0.1 s : 1.0e-9 > 0.02 s : 6.5e-9 > > Counter B > Gate time : STDDEV > 1 s : 1.3--9 > 0.1 s : 1.5e-8 > 0.02 s : 1.4e-7 > > The results have been verified by performing multiple measurements. Counter > A and B are both fractional counters that use interpolation. > > The manual of the Agilent 53132A specifies the worst case RMS error of a > frequency measurement for different gate times and an input frequency of > 10MHz as: > > Agilent 53132A > Gate time : Max RMS error (estimated) > 1 s : 2e-10 > 0.1 s : 2e-9 > 0.02 s : 5e-8 > > Assuming the RMS error and the STDDEV are the same the steps with gate time > change of the Agilent and Counter A seem to be comparable but Counter B > behaves a bit different for 0.02 s gate time. > > This leads me to the following questions: > Is measuring the STDDEV of the ratio of two input frequencies derived from > the same timebase a valid way to assess the short term measurement accuracy > of a frequency counter? > If not, how should this be done? > If yes, do the numbers I'v listed above make sense? > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
TV
Tom Van Baak
Mon, Jan 24, 2022 4:26 PM

Erik,

See some plots of testing a 53132A counter in time interval mode:

http://leapsecond.com/pages/53132/

Yes, using stdev (rms) is ok. You can also use ADEV and expect to see a
line with slope of -1; the noise of the counter is the value at tau 0
[*]. You can also use TDEV and it should be a relatively flat line; this
also reflects the noise floor of the counter. If your stdev varies a lot
depending on sample size, or if your ADEV plot isn't a very straight
line with slope exactly -1, there's a problem.

To determine the noise floor of a counter it's best to use different
sources, independent in frequency and also independent in phase. That
rules out dividers or multipliers. The phase needs to "sweep" across a
wide range. What you want to avoid are "sweet spots" where there is some
constant relationship between the counter's internal reference and the
external input(s) that are being measured. A sweep will help identify
any subtle frequency or phase pulling effects. You can see this
dramatically in the 53132A plots.

So in frequency mode make sure it's two different oscillators. In time
interval mode make sure it's 3 independent sources; all drifting in
phase wrt to each other. You're welcome to use only one source too but
that will likely give an overly optimistic report of counter resolution.

An analogy is if you built a 5 digit 20 volt DIY voltmeter and only
tested it at 0.0000 volts, or maybe only at 12.345 volts. Sure, you can
take hundreds of readings and computer the stdev. But a better way is to
generate a precise triangle wave from 0 to 20 volts in steps of 20 uV
and then look at the readings to see how accurate they are across the
entire range.

Most frequency counters have artifacts that cause some level of
non-linearity depending on where the input(s) land wrt to the reference.
It could be minor, but it's something that contributes to the accuracy
and noise floor, especially with interpolating or "enhanced resolution"
counter. It's ok. You just want to know how good or bad it is. It can
even be a diagnostic to find PCB layout or f/w bugs.

The 10 MHz effect that Bob mentions is:

http://leapsecond.com/pages/53132/53132-reduced-resolution.gif

/tvb

On 1/24/2022 6:47 AM, Bob kb8tq wrote:

Hi

First off, yes, standard deviation is a pretty good way to look at what a counter
is doing. Reducing the answer to time ( = picoseconds ) is usually the easy way
to look at the data.

Next up, counters have a lot of things that impact what they do. The slew rate
of the input signal is a big one on most counters ( = square waves and sine
waves will give you different answers).  It is best to do your testing with the
type of signal you are most likely to use.

Some counters do very odd things when the input and reference are tightly coupled.
An SR620 counting it’s reference output is one good example of this. Best practice
is to run two independent sources. One supplies the reference, a second one
generates the test signal.

HP counters (and likely some others) have interesting “dead spots” at 10 MHz, at
10MHz / N and 10 MHz * N. to get the best performance numbers, test with a signal
that is not in one of these regions.

It is worth looking at the ADEV / phase noise of your test signal. You can indeed
get into trouble there…..

Fun !!!

Bob

On Jan 24, 2022, at 8:43 AM, Erik Kaashoek erik@kaashoek.com wrote:

For some project I'm trying to establish the short term accuracy of a
frequency counter versus the gate time.
As using the Allan Deviation for this type of measurement did lead to
extensive discussion over the validity of using ADEV for measuring the
short term performance of a counter I tried to find a different, but still
relevant way to establish the performance.
To exclude as much as possible external and long term factors I'm using a
single fairly stable OCXO (short term error below 1e-10) to output 10 MHz.
This 10 MHz goes into an SI5351 as reference for its PLL and the SI5351
outputs two frequencies from the same VCO, one at 10 MHz into input A of
the counter and one at 10.00003319 MHz into input B of the counter. The
counter is setup to measure the ratio of A/B and to display the STDDEV of
the ratio over n=100. The STDEV of counter B is calculated as the square
root of  ( (the sum of the squares of the difference between the measured
ratio and the average ratio ) divided by the number of measurements )
I'm aware the SI5351 uses a fractional divider but I hope the impact is
below the measurement accuracy required.
Doing this test with two counter gave these results:

Counter A
Gate time :  STDDEV
1 s : 1.0-10
0.1 s : 1.0e-9
0.02 s : 6.5e-9

Counter B
Gate time :  STDDEV
1 s : 1.3--9
0.1 s : 1.5e-8
0.02 s : 1.4e-7

The results have been verified by performing multiple measurements. Counter
A and B are both fractional counters that use interpolation.

The manual of the Agilent 53132A specifies the worst case RMS  error of a
frequency measurement for different gate times and an input frequency of
10MHz as:

Agilent 53132A
Gate time : Max RMS error (estimated)
1 s :  2e-10
0.1 s : 2e-9
0.02 s :  5e-8

Assuming the RMS error and the STDDEV are the same the steps with gate time
change of the Agilent and Counter A seem to be comparable but Counter B
behaves a bit different for 0.02 s gate time.

This leads me to the following questions:
Is measuring the STDDEV of the ratio of two input frequencies derived from
the same timebase a valid way to assess the short term measurement accuracy
of a frequency counter?
If not, how should this be done?
If yes, do the numbers I'v listed above make sense?


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.


time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Erik, See some plots of testing a 53132A counter in time interval mode: http://leapsecond.com/pages/53132/ Yes, using stdev (rms) is ok. You can also use ADEV and expect to see a line with slope of -1; the noise of the counter is the value at tau 0 [*]. You can also use TDEV and it should be a relatively flat line; this also reflects the noise floor of the counter. If your stdev varies a lot depending on sample size, or if your ADEV plot isn't a very straight line with slope exactly -1, there's a problem. To determine the noise floor of a counter it's best to use different sources, independent in frequency and also independent in phase. That rules out dividers or multipliers. The phase needs to "sweep" across a wide range. What you want to avoid are "sweet spots" where there is some constant relationship between the counter's internal reference and the external input(s) that are being measured. A sweep will help identify any subtle frequency or phase pulling effects. You can see this dramatically in the 53132A plots. So in frequency mode make sure it's two different oscillators. In time interval mode make sure it's 3 independent sources; all drifting in phase wrt to each other. You're welcome to use only one source too but that will likely give an overly optimistic report of counter resolution. An analogy is if you built a 5 digit 20 volt DIY voltmeter and only tested it at 0.0000 volts, or maybe only at 12.345 volts. Sure, you can take hundreds of readings and computer the stdev. But a better way is to generate a precise triangle wave from 0 to 20 volts in steps of 20 uV and then look at the readings to see how accurate they are across the entire range. Most frequency counters have artifacts that cause some level of non-linearity depending on where the input(s) land wrt to the reference. It could be minor, but it's something that contributes to the accuracy and noise floor, especially with interpolating or "enhanced resolution" counter. It's ok. You just want to know how good or bad it is. It can even be a diagnostic to find PCB layout or f/w bugs. The 10 MHz effect that Bob mentions is: http://leapsecond.com/pages/53132/53132-reduced-resolution.gif /tvb On 1/24/2022 6:47 AM, Bob kb8tq wrote: > Hi > > First off, yes, standard deviation is a pretty good way to look at what a counter > is doing. Reducing the answer to time ( = picoseconds ) is usually the easy way > to look at the data. > > Next up, counters have a *lot* of things that impact what they do. The slew rate > of the input signal is a big one on most counters ( = square waves and sine > waves will give you different answers). It is best to do your testing with the > type of signal you are most likely to use. > > Some counters do very odd things when the input and reference are tightly coupled. > An SR620 counting it’s reference output is one good example of this. Best practice > is to run two independent sources. One supplies the reference, a second one > generates the test signal. > > HP counters (and likely some others) have interesting “dead spots” at 10 MHz, at > 10MHz / N and 10 MHz * N. to get the best performance numbers, test with a signal > that is not in one of these regions. > > It is worth looking at the ADEV / phase noise of your test signal. You can indeed > get into trouble there….. > > Fun !!! > > Bob > >> On Jan 24, 2022, at 8:43 AM, Erik Kaashoek <erik@kaashoek.com> wrote: >> >> For some project I'm trying to establish the short term accuracy of a >> frequency counter versus the gate time. >> As using the Allan Deviation for this type of measurement did lead to >> extensive discussion over the validity of using ADEV for measuring the >> short term performance of a counter I tried to find a different, but still >> relevant way to establish the performance. >> To exclude as much as possible external and long term factors I'm using a >> single fairly stable OCXO (short term error below 1e-10) to output 10 MHz. >> This 10 MHz goes into an SI5351 as reference for its PLL and the SI5351 >> outputs two frequencies from the same VCO, one at 10 MHz into input A of >> the counter and one at 10.00003319 MHz into input B of the counter. The >> counter is setup to measure the ratio of A/B and to display the STDDEV of >> the ratio over n=100. The STDEV of counter B is calculated as the square >> root of ( (the sum of the squares of the difference between the measured >> ratio and the average ratio ) divided by the number of measurements ) >> I'm aware the SI5351 uses a fractional divider but I hope the impact is >> below the measurement accuracy required. >> Doing this test with two counter gave these results: >> >> Counter A >> Gate time : STDDEV >> 1 s : 1.0-10 >> 0.1 s : 1.0e-9 >> 0.02 s : 6.5e-9 >> >> Counter B >> Gate time : STDDEV >> 1 s : 1.3--9 >> 0.1 s : 1.5e-8 >> 0.02 s : 1.4e-7 >> >> The results have been verified by performing multiple measurements. Counter >> A and B are both fractional counters that use interpolation. >> >> The manual of the Agilent 53132A specifies the worst case RMS error of a >> frequency measurement for different gate times and an input frequency of >> 10MHz as: >> >> Agilent 53132A >> Gate time : Max RMS error (estimated) >> 1 s : 2e-10 >> 0.1 s : 2e-9 >> 0.02 s : 5e-8 >> >> Assuming the RMS error and the STDDEV are the same the steps with gate time >> change of the Agilent and Counter A seem to be comparable but Counter B >> behaves a bit different for 0.02 s gate time. >> >> This leads me to the following questions: >> Is measuring the STDDEV of the ratio of two input frequencies derived from >> the same timebase a valid way to assess the short term measurement accuracy >> of a frequency counter? >> If not, how should this be done? >> If yes, do the numbers I'v listed above make sense? >> _______________________________________________ >> time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com >> To unsubscribe, go to and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an email to time-nuts-leave@lists.febo.com > To unsubscribe, go to and follow the instructions there.
PK
Poul-Henning Kamp
Mon, Jan 24, 2022 4:44 PM

Tom Van Baak writes:

To determine the noise floor of a counter it's best to use different
sources, independent in frequency and also independent in phase. That
rules out dividers or multipliers. The phase needs to "sweep" across a
wide range.

A neat way to do this:

Clock --+-- (ext timebase) HP3336 ----->
        |                                  TI-counter
        +------------------------------>

Set the HP3336 to the same frequency as the clock source
and use it to vary the relative phase of the two signals
into the counter.

The main advantage is that you can make many measurements at each
of the 360 phase differences and thus get both avg+stddev along
the curve.

Any unlinearities in the HP3336 obviously mix into the resulting
measurements, but as far as I can tell, it performs much better
than any of my counters.

That is probably not unrelated to it being built to measure "time
distortion" on telephony channels frequency-stacked on coax cables :-)

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- Tom Van Baak writes: > To determine the noise floor of a counter it's best to use different > sources, independent in frequency and also independent in phase. That > rules out dividers or multipliers. The phase needs to "sweep" across a > wide range. A neat way to do this: Clock --+-- (ext timebase) HP3336 -----> | TI-counter +------------------------------> Set the HP3336 to the same frequency as the clock source and use it to vary the relative phase of the two signals into the counter. The main advantage is that you can make many measurements at each of the 360 phase differences and thus get both avg+stddev along the curve. Any unlinearities in the HP3336 obviously mix into the resulting measurements, but as far as I can tell, it performs much better than any of my counters. That is probably not unrelated to it being built to measure "time distortion" on telephony channels frequency-stacked on coax cables :-) -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.