In an effort to get a 'standard' for DC and AC voltage and current, along
with resistance, I recently acquired a Fluke 5100B.

After an effort, finding several dead/shorted/blown (literally) tantalum
capacitors as well as shorted zeners and loose mounting screws for the HV
output section transistors, it seems to be working on all ranges.  It has
the wideband frequency option (03) and the HPIB option (05) but I have not
had a chance to try those out yet.

Anyone on the list with any experience with this unit?  Any hints,
suggestions, pitfalls I need to know about?

I downloaded the manual from the Fluke site.

My plan is to use it along with an Agilent calibrated 3458A to be able to
calibrate other DMM's.

Any suggestions appreciated.

Thanks in advance.

Joe

In an effort to get a 'standard' for DC and AC voltage and current, along with resistance, I recently acquired a Fluke 5100B. After an effort, finding several dead/shorted/blown (literally) tantalum capacitors as well as shorted zeners and loose mounting screws for the HV output section transistors, it seems to be working on all ranges. It has the wideband frequency option (03) and the HPIB option (05) but I have not had a chance to try those out yet. Anyone on the list with any experience with this unit? Any hints, suggestions, pitfalls I need to know about? I downloaded the manual from the Fluke site. My plan is to use it along with an Agilent calibrated 3458A to be able to calibrate other DMM's. Any suggestions appreciated. Thanks in advance. Joe
PL
Pete Lancashire
Sun, Sep 2, 2012 3:46 AM

I would now go after all the tants. They have the bad habit of doing the
power supply in.

Another warning, never power the unit up if it has not been bottom down
for at least a couple hours. The wonderful world of Hg wetted relays.

On Sat, Sep 1, 2012 at 7:25 PM, J. L. Trantham jltran@att.net wrote:

In an effort to get a 'standard' for DC and AC voltage and current, along
with resistance, I recently acquired a Fluke 5100B.

After an effort, finding several dead/shorted/blown (literally) tantalum
capacitors as well as shorted zeners and loose mounting screws for the HV
output section transistors, it seems to be working on all ranges.  It has
the wideband frequency option (03) and the HPIB option (05) but I have not
had a chance to try those out yet.

Anyone on the list with any experience with this unit?  Any hints,
suggestions, pitfalls I need to know about?

I downloaded the manual from the Fluke site.

My plan is to use it along with an Agilent calibrated 3458A to be able to
calibrate other DMM's.

Any suggestions appreciated.

Thanks in advance.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I would now go after all the tants. They have the bad habit of doing the power supply in. Another warning, never power the unit up if it has not been bottom down for at least a couple hours. The wonderful world of Hg wetted relays. On Sat, Sep 1, 2012 at 7:25 PM, J. L. Trantham <jltran@att.net> wrote: > In an effort to get a 'standard' for DC and AC voltage and current, along > with resistance, I recently acquired a Fluke 5100B. > > > > After an effort, finding several dead/shorted/blown (literally) tantalum > capacitors as well as shorted zeners and loose mounting screws for the HV > output section transistors, it seems to be working on all ranges. It has > the wideband frequency option (03) and the HPIB option (05) but I have not > had a chance to try those out yet. > > > > Anyone on the list with any experience with this unit? Any hints, > suggestions, pitfalls I need to know about? > > > > I downloaded the manual from the Fluke site. > > > > My plan is to use it along with an Agilent calibrated 3458A to be able to > calibrate other DMM's. > > > > Any suggestions appreciated. > > > > Thanks in advance. > > > > Joe > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
JL
J. L. Trantham
Sun, Sep 2, 2012 11:58 AM

Pete,

Thanks for the reply.

There were two tantalums on the HV board that were, literally, exploded.
Just an empty shell.  I could see no other obvious issues except one of the
power transistors on the A17A1 boards was 'loose', one mounting screw not
screwed in all the way and the other 'tight' but with crossed threads.  I
replaced the two obvious tantalum problems, properly installed the power
transistor and all power supplies came up except the -15 VDC, sitting around

  • 0.5 VDC.  I found the -15 VDC regulator very hot, two series 6.8 V zeners
    used to drop the -40 VDC unreg to about -26 VDC before it got to the LM7915
    were shorted and the resistance on the -15 VDC buss at about 0.5 ohm to
    ground.

I didn't have any 6.8 V zeners so I installed three 5.1 V zeners and started
lifting cards.  The resistance on the -15 VDC buss went to about 600 ohms by
lifting the DAC board.  Found a shorted tantalum with no obvious external
damage, fired it up and, again, the -15 VDC buss would not come up.  Again,
about 0.5 ohm to ground and this time the shorted tantalum was on the Analog
Control board.  Replaced it and now, it all seems OK.

The three 'new' zeners and LM7915 all seemed to survive the 'new' shorted
tantalum without a problem though they were, again, quite hot when I removed
the A9 Power Supply Regulator Board.

My A9 board is different from the schematic in the manual.  My board uses an
LM7915 with two 6.8 V zeners in series to the - 40 VDC input to supply - 15
VDC where the schematic and parts list show a LM7905 with a 10 V zener in
the ground lead and a single 6.8 V zener in line with -40 VDC supply.

I think I have seen three different part numbers for manuals.  I ordered one
from theBay that is not here yet.  Any other sources of manuals on the net
that you know of?  I would like to find one (or at least the 'change
sheets') that apply to my serial number.

I had read about the 5102B, a 5100B (or is it a 5101B) that comes in a
'case' and is designed for any/all position operation, not just sitting
upright.  I didn't know about the 'pre-flight' upright position time
requirement on the 5100B.  It looks like I got away with it this time
though.

Thanks again for your reply and help.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Pete Lancashire
Sent: Saturday, September 01, 2012 10:46 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 5100B Calibrator

I would now go after all the tants. They have the bad habit of doing the
power supply in.

Another warning, never power the unit up if it has not been bottom down for
at least a couple hours. The wonderful world of Hg wetted relays.

On Sat, Sep 1, 2012 at 7:25 PM, J. L. Trantham jltran@att.net wrote:

In an effort to get a 'standard' for DC and AC voltage and current,
along with resistance, I recently acquired a Fluke 5100B.

After an effort, finding several dead/shorted/blown (literally)
tantalum capacitors as well as shorted zeners and loose mounting
screws for the HV output section transistors, it seems to be working
on all ranges.  It has the wideband frequency option (03) and the HPIB
option (05) but I have not had a chance to try those out yet.

Anyone on the list with any experience with this unit?  Any hints,
suggestions, pitfalls I need to know about?

I downloaded the manual from the Fluke site.

My plan is to use it along with an Agilent calibrated 3458A to be able
to calibrate other DMM's.

Any suggestions appreciated.

Thanks in advance.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Pete, Thanks for the reply. There were two tantalums on the HV board that were, literally, exploded. Just an empty shell. I could see no other obvious issues except one of the power transistors on the A17A1 boards was 'loose', one mounting screw not screwed in all the way and the other 'tight' but with crossed threads. I replaced the two obvious tantalum problems, properly installed the power transistor and all power supplies came up except the -15 VDC, sitting around - 0.5 VDC. I found the -15 VDC regulator very hot, two series 6.8 V zeners used to drop the -40 VDC unreg to about -26 VDC before it got to the LM7915 were shorted and the resistance on the -15 VDC buss at about 0.5 ohm to ground. I didn't have any 6.8 V zeners so I installed three 5.1 V zeners and started lifting cards. The resistance on the -15 VDC buss went to about 600 ohms by lifting the DAC board. Found a shorted tantalum with no obvious external damage, fired it up and, again, the -15 VDC buss would not come up. Again, about 0.5 ohm to ground and this time the shorted tantalum was on the Analog Control board. Replaced it and now, it all seems OK. The three 'new' zeners and LM7915 all seemed to survive the 'new' shorted tantalum without a problem though they were, again, quite hot when I removed the A9 Power Supply Regulator Board. My A9 board is different from the schematic in the manual. My board uses an LM7915 with two 6.8 V zeners in series to the - 40 VDC input to supply - 15 VDC where the schematic and parts list show a LM7905 with a 10 V zener in the ground lead and a single 6.8 V zener in line with -40 VDC supply. I think I have seen three different part numbers for manuals. I ordered one from theBay that is not here yet. Any other sources of manuals on the net that you know of? I would like to find one (or at least the 'change sheets') that apply to my serial number. I had read about the 5102B, a 5100B (or is it a 5101B) that comes in a 'case' and is designed for any/all position operation, not just sitting upright. I didn't know about the 'pre-flight' upright position time requirement on the 5100B. It looks like I got away with it this time though. Thanks again for your reply and help. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Pete Lancashire Sent: Saturday, September 01, 2012 10:46 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 5100B Calibrator I would now go after all the tants. They have the bad habit of doing the power supply in. Another warning, never power the unit up if it has not been bottom down for at least a couple hours. The wonderful world of Hg wetted relays. On Sat, Sep 1, 2012 at 7:25 PM, J. L. Trantham <jltran@att.net> wrote: > In an effort to get a 'standard' for DC and AC voltage and current, > along with resistance, I recently acquired a Fluke 5100B. > > > > After an effort, finding several dead/shorted/blown (literally) > tantalum capacitors as well as shorted zeners and loose mounting > screws for the HV output section transistors, it seems to be working > on all ranges. It has the wideband frequency option (03) and the HPIB > option (05) but I have not had a chance to try those out yet. > > > > Anyone on the list with any experience with this unit? Any hints, > suggestions, pitfalls I need to know about? > > > > I downloaded the manual from the Fluke site. > > > > My plan is to use it along with an Agilent calibrated 3458A to be able > to calibrate other DMM's. > > > > Any suggestions appreciated. > > > > Thanks in advance. > > > > Joe > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.