It is well known that by changing the nominal operating current of a 1N825 type reference diode, that its voltage turn over temperature point can be moved.
Usually the turn over temperature can be set to the nominal operating temperature,  there by giving a Zero TC reference voltage over a limited temperature range.
For the parts I'm testing, the voltage changes less than 0.1 PPM with a 5 deg F temperature variation, and 1PPM over a 20 deg F range when the zero TC current is set for room temperature.
Not bad, but being an extreme Volt Nut, I always want to do better.

Two ways to make it better, Add an X^2 (second order) temperature compensator or put the reference in an oven.

What I did was make a very simple Mini-Oven using an 1N825, a glass thermister & a 1/4 resistor,
all heat shrink together  in a small package about the size of a 1 watt resistor.
Now by Appling about 1/8 W nominal power to the resistor from a simple temperature controller, the thermistor can be held at a constant Temperature.
This allows setting it up so that normal temperature changes have true ZERO effect on the zero TCed zener voltage.
Next thing I want to make is a simple times 8/5 digital gain amp so that the gain resistors will have no effect when amplifying the 6.25x volt zener  up to 10.000000 Volts.
(or times 10/7 for use with other types of voltage references)

ws

It is well known that by changing the nominal operating current of a 1N825 type reference diode, that its voltage turn over temperature point can be moved. Usually the turn over temperature can be set to the nominal operating temperature, there by giving a Zero TC reference voltage over a limited temperature range. For the parts I'm testing, the voltage changes less than 0.1 PPM with a 5 deg F temperature variation, and 1PPM over a 20 deg F range when the zero TC current is set for room temperature. Not bad, but being an extreme Volt Nut, I always want to do better. Two ways to make it better, Add an X^2 (second order) temperature compensator or put the reference in an oven. What I did was make a very simple Mini-Oven using an 1N825, a glass thermister & a 1/4 resistor, all heat shrink together in a small package about the size of a 1 watt resistor. Now by Appling about 1/8 W nominal power to the resistor from a simple temperature controller, the thermistor can be held at a constant Temperature. This allows setting it up so that normal temperature changes have true ZERO effect on the zero TCed zener voltage. Next thing I want to make is a simple times 8/5 digital gain amp so that the gain resistors will have no effect when amplifying the 6.25x volt zener up to 10.000000 Volts. (or times 10/7 for use with other types of voltage references) ws
MK
m k
Thu, Sep 15, 2011 9:25 PM

Hi Warren,

That does seem a very simple way to get the diode at a steady temp, but then you need to keep the rest steady to stop all the other uncertainties swamping the nice steady diode you have just created.

From: warrensjmail-one@yahoo.com
To: volt-nuts@febo.com
Date: Thu, 15 Sep 2011 14:15:58 -0700
Subject: [volt-nuts] Voltage Reference Mini-Oven

It is well known that by changing the nominal operating current of a 1N825 type reference diode, that its voltage turn over temperature point can be moved.
Usually the turn over temperature can be set to the nominal operating temperature,  there by giving a Zero TC reference voltage over a limited temperature range.
For the parts I'm testing, the voltage changes less than 0.1 PPM with a 5 deg F temperature variation, and 1PPM over a 20 deg F range when the zero TC current is set for room temperature.
Not bad, but being an extreme Volt Nut, I always want to do better.

Two ways to make it better, Add an X^2 (second order) temperature compensator or put the reference in an oven.

What I did was make a very simple Mini-Oven using an 1N825, a glass thermister & a 1/4 resistor,
all heat shrink together  in a small package about the size of a 1 watt resistor.
Now by Appling about 1/8 W nominal power to the resistor from a simple temperature controller, the thermistor can be held at a constant Temperature.
This allows setting it up so that normal temperature changes have true ZERO effect on the zero TCed zener voltage.
Next thing I want to make is a simple times 8/5 digital gain amp so that the gain resistors will have no effect when amplifying the 6.25x volt zener  up to 10.000000 Volts.
(or times 10/7 for use with other types of voltage references)

ws


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Warren, That does seem a very simple way to get the diode at a steady temp, but then you need to keep the rest steady to stop all the other uncertainties swamping the nice steady diode you have just created. > From: warrensjmail-one@yahoo.com > To: volt-nuts@febo.com > Date: Thu, 15 Sep 2011 14:15:58 -0700 > Subject: [volt-nuts] Voltage Reference Mini-Oven > > > It is well known that by changing the nominal operating current of a 1N825 type reference diode, that its voltage turn over temperature point can be moved. > Usually the turn over temperature can be set to the nominal operating temperature, there by giving a Zero TC reference voltage over a limited temperature range. > For the parts I'm testing, the voltage changes less than 0.1 PPM with a 5 deg F temperature variation, and 1PPM over a 20 deg F range when the zero TC current is set for room temperature. > Not bad, but being an extreme Volt Nut, I always want to do better. > > Two ways to make it better, Add an X^2 (second order) temperature compensator or put the reference in an oven. > > What I did was make a very simple Mini-Oven using an 1N825, a glass thermister & a 1/4 resistor, > all heat shrink together in a small package about the size of a 1 watt resistor. > Now by Appling about 1/8 W nominal power to the resistor from a simple temperature controller, the thermistor can be held at a constant Temperature. > This allows setting it up so that normal temperature changes have true ZERO effect on the zero TCed zener voltage. > Next thing I want to make is a simple times 8/5 digital gain amp so that the gain resistors will have no effect when amplifying the 6.25x volt zener up to 10.000000 Volts. > (or times 10/7 for use with other types of voltage references) > > ws > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
Z
zbigniew169
Thu, Sep 15, 2011 10:03 PM

He I am Zbigniew W. Kaminski from Warsaw Poland. I was member-observer up to
this time. I am user of Prema 6040D 7 & 1/2 digits multimeter.

I would be appreciate if You will send some details concerning construction
of Your mini-oven. I need something similar for Weston saturated standard
cell. In my case the stability must be even better than 0.1 Celsjus degree,
because of Weston temco in the range about 50 ppm. I think that in my
project I should use patinium resistor rathen than termistor, because it's
much long term stable than termistor. Platinium element is much less
sensitive to temperature, therefore the cuircuit  will be much complicated.
I am an owner of single Weston cell which stays in my lab during last 20
years, and I am curious how stable it is.

2011/9/15 WarrenS warrensjmail-one@yahoo.com

It is well known that by changing the nominal operating current of a 1N825
type reference diode, that its voltage turn over temperature point can be
moved.
Usually the turn over temperature can be set to the nominal operating
temperature,  there by giving a Zero TC reference voltage over a limited
temperature range.
For the parts I'm testing, the voltage changes less than 0.1 PPM with a 5
deg F temperature variation, and 1PPM over a 20 deg F range when the zero TC
current is set for room temperature.
Not bad, but being an extreme Volt Nut, I always want to do better.

Two ways to make it better, Add an X^2 (second order) temperature
compensator or put the reference in an oven.

What I did was make a very simple Mini-Oven using an 1N825, a glass
thermister & a 1/4 resistor,
all heat shrink together  in a small package about the size of a 1 watt
resistor.
Now by Appling about 1/8 W nominal power to the resistor from a simple
temperature controller, the thermistor can be held at a constant
Temperature.
This allows setting it up so that normal temperature changes have true ZERO
effect on the zero TCed zener voltage.
Next thing I want to make is a simple times 8/5 digital gain amp so that
the gain resistors will have no effect when amplifying the 6.25x volt zener
up to 10.000000 Volts.
(or times 10/7 for use with other types of voltage references)

ws


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Zbyszek

He I am Zbigniew W. Kaminski from Warsaw Poland. I was member-observer up to this time. I am user of Prema 6040D 7 & 1/2 digits multimeter. I would be appreciate if You will send some details concerning construction of Your mini-oven. I need something similar for Weston saturated standard cell. In my case the stability must be even better than 0.1 Celsjus degree, because of Weston temco in the range about 50 ppm. I think that in my project I should use patinium resistor rathen than termistor, because it's much long term stable than termistor. Platinium element is much less sensitive to temperature, therefore the cuircuit will be much complicated. I am an owner of single Weston cell which stays in my lab during last 20 years, and I am curious how stable it is. 2011/9/15 WarrenS <warrensjmail-one@yahoo.com> > > It is well known that by changing the nominal operating current of a 1N825 > type reference diode, that its voltage turn over temperature point can be > moved. > Usually the turn over temperature can be set to the nominal operating > temperature, there by giving a Zero TC reference voltage over a limited > temperature range. > For the parts I'm testing, the voltage changes less than 0.1 PPM with a 5 > deg F temperature variation, and 1PPM over a 20 deg F range when the zero TC > current is set for room temperature. > Not bad, but being an extreme Volt Nut, I always want to do better. > > Two ways to make it better, Add an X^2 (second order) temperature > compensator or put the reference in an oven. > > What I did was make a very simple Mini-Oven using an 1N825, a glass > thermister & a 1/4 resistor, > all heat shrink together in a small package about the size of a 1 watt > resistor. > Now by Appling about 1/8 W nominal power to the resistor from a simple > temperature controller, the thermistor can be held at a constant > Temperature. > This allows setting it up so that normal temperature changes have true ZERO > effect on the zero TCed zener voltage. > Next thing I want to make is a simple times 8/5 digital gain amp so that > the gain resistors will have no effect when amplifying the 6.25x volt zener > up to 10.000000 Volts. > (or times 10/7 for use with other types of voltage references) > > ws > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > -- Zbyszek
FS
Fred Schneider
Thu, Sep 15, 2011 10:30 PM

I have the original guildline enclosure with schematic. There is an oven in there and the circuit is rather straight forward using some tranistors. Maybe you can use this as an example. The saturated cells must be hold at 30 degrees. They must kept vertical allways and viberation free. After moving them arround they need about two weeks to stabilize and do not use a 10M meter on them. Mine are from 1971 and still in very good shape.

Fred PA4TIM

Op 16 sep. 2011 om 00:03 heeft zbigniew169 zbigniew169@gmail.com het volgende geschreven:

He I am Zbigniew W. Kaminski from Warsaw Poland. I was member-observer up to
this time. I am user of Prema 6040D 7 & 1/2 digits multimeter.

I would be appreciate if You will send some details concerning construction
of Your mini-oven. I need something similar for Weston saturated standard
cell. In my case the stability must be even better than 0.1 Celsjus degree,
because of Weston temco in the range about 50 ppm. I think that in my
project I should use patinium resistor rathen than termistor, because it's
much long term stable than termistor. Platinium element is much less
sensitive to temperature, therefore the cuircuit  will be much complicated.
I am an owner of single Weston cell which stays in my lab during last 20
years, and I am curious how stable it is.

2011/9/15 WarrenS warrensjmail-one@yahoo.com

It is well known that by changing the nominal operating current of a 1N825
type reference diode, that its voltage turn over temperature point can be
moved.
Usually the turn over temperature can be set to the nominal operating
temperature,  there by giving a Zero TC reference voltage over a limited
temperature range.
For the parts I'm testing, the voltage changes less than 0.1 PPM with a 5
deg F temperature variation, and 1PPM over a 20 deg F range when the zero TC
current is set for room temperature.
Not bad, but being an extreme Volt Nut, I always want to do better.

Two ways to make it better, Add an X^2 (second order) temperature
compensator or put the reference in an oven.

What I did was make a very simple Mini-Oven using an 1N825, a glass
thermister & a 1/4 resistor,
all heat shrink together  in a small package about the size of a 1 watt
resistor.
Now by Appling about 1/8 W nominal power to the resistor from a simple
temperature controller, the thermistor can be held at a constant
Temperature.
This allows setting it up so that normal temperature changes have true ZERO
effect on the zero TCed zener voltage.
Next thing I want to make is a simple times 8/5 digital gain amp so that
the gain resistors will have no effect when amplifying the 6.25x volt zener
up to 10.000000 Volts.
(or times 10/7 for use with other types of voltage references)

ws


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Zbyszek


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I have the original guildline enclosure with schematic. There is an oven in there and the circuit is rather straight forward using some tranistors. Maybe you can use this as an example. The saturated cells must be hold at 30 degrees. They must kept vertical allways and viberation free. After moving them arround they need about two weeks to stabilize and do not use a 10M meter on them. Mine are from 1971 and still in very good shape. Fred PA4TIM Op 16 sep. 2011 om 00:03 heeft zbigniew169 <zbigniew169@gmail.com> het volgende geschreven: > He I am Zbigniew W. Kaminski from Warsaw Poland. I was member-observer up to > this time. I am user of Prema 6040D 7 & 1/2 digits multimeter. > > I would be appreciate if You will send some details concerning construction > of Your mini-oven. I need something similar for Weston saturated standard > cell. In my case the stability must be even better than 0.1 Celsjus degree, > because of Weston temco in the range about 50 ppm. I think that in my > project I should use patinium resistor rathen than termistor, because it's > much long term stable than termistor. Platinium element is much less > sensitive to temperature, therefore the cuircuit will be much complicated. > I am an owner of single Weston cell which stays in my lab during last 20 > years, and I am curious how stable it is. > > 2011/9/15 WarrenS <warrensjmail-one@yahoo.com> > >> >> It is well known that by changing the nominal operating current of a 1N825 >> type reference diode, that its voltage turn over temperature point can be >> moved. >> Usually the turn over temperature can be set to the nominal operating >> temperature, there by giving a Zero TC reference voltage over a limited >> temperature range. >> For the parts I'm testing, the voltage changes less than 0.1 PPM with a 5 >> deg F temperature variation, and 1PPM over a 20 deg F range when the zero TC >> current is set for room temperature. >> Not bad, but being an extreme Volt Nut, I always want to do better. >> >> Two ways to make it better, Add an X^2 (second order) temperature >> compensator or put the reference in an oven. >> >> What I did was make a very simple Mini-Oven using an 1N825, a glass >> thermister & a 1/4 resistor, >> all heat shrink together in a small package about the size of a 1 watt >> resistor. >> Now by Appling about 1/8 W nominal power to the resistor from a simple >> temperature controller, the thermistor can be held at a constant >> Temperature. >> This allows setting it up so that normal temperature changes have true ZERO >> effect on the zero TCed zener voltage. >> Next thing I want to make is a simple times 8/5 digital gain amp so that >> the gain resistors will have no effect when amplifying the 6.25x volt zener >> up to 10.000000 Volts. >> (or times 10/7 for use with other types of voltage references) >> >> ws >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> > > > > -- > Zbyszek > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
Z
zbigniew169
Thu, Sep 15, 2011 11:21 PM

I would e grateful for a copy of such documentation.

By the way; Weston saturated standard cells are still in use in our polish
equivalent of american NBS; they kept about 400 units in Glowny Urzad Miar
in Warsaw.
In 1937 modified Weston cell was build it has 11% Cd 15% Bi and remain Hg as
amalgamate. It had 13 ppm tempo almoust 5 time less than original Weston
cell. In 1964 Roger G. Bates wrote a paper regarding stability o a such
secondary standard; the EFM of such cell was 1.0188 V at 25 Cel.deg., Six
cells after 26 years changed their EMF average by 0.03 mV. Because I am
biochemist, so electrochemisty was slightly close to my interest. Sometimes
I am dreaming to build such a cell...

Best regards, from Warsaw,

2011/9/16 Fred Schneider pa4tim@gmail.com

I have the original guildline enclosure with schematic. There is an oven in
there and the circuit is rather straight forward using some tranistors.
Maybe you can use this as an example. The saturated cells must be hold at 30
degrees. They must kept vertical allways and viberation free. After moving
them arround they need about two weeks to stabilize and do not use a 10M
meter on them. Mine are from 1971 and still in very good shape.

Fred PA4TIM

Op 16 sep. 2011 om 00:03 heeft zbigniew169 zbigniew169@gmail.com het
volgende geschreven:

He I am Zbigniew W. Kaminski from Warsaw Poland. I was member-observer up

to

this time. I am user of Prema 6040D 7 & 1/2 digits multimeter.

I would be appreciate if You will send some details concerning

construction

of Your mini-oven. I need something similar for Weston saturated standard
cell. In my case the stability must be even better than 0.1 Celsjus

degree,

because of Weston temco in the range about 50 ppm. I think that in my
project I should use patinium resistor rathen than termistor, because

it's

much long term stable than termistor. Platinium element is much less
sensitive to temperature, therefore the cuircuit  will be much

complicated.

I am an owner of single Weston cell which stays in my lab during last 20
years, and I am curious how stable it is.

2011/9/15 WarrenS warrensjmail-one@yahoo.com

It is well known that by changing the nominal operating current of a

1N825

type reference diode, that its voltage turn over temperature point can

be

moved.
Usually the turn over temperature can be set to the nominal operating
temperature,  there by giving a Zero TC reference voltage over a limited
temperature range.
For the parts I'm testing, the voltage changes less than 0.1 PPM with a

5

deg F temperature variation, and 1PPM over a 20 deg F range when the

zero TC

current is set for room temperature.
Not bad, but being an extreme Volt Nut, I always want to do better.

Two ways to make it better, Add an X^2 (second order) temperature
compensator or put the reference in an oven.

What I did was make a very simple Mini-Oven using an 1N825, a glass
thermister & a 1/4 resistor,
all heat shrink together  in a small package about the size of a 1 watt
resistor.
Now by Appling about 1/8 W nominal power to the resistor from a simple
temperature controller, the thermistor can be held at a constant
Temperature.
This allows setting it up so that normal temperature changes have true

ZERO

effect on the zero TCed zener voltage.
Next thing I want to make is a simple times 8/5 digital gain amp so that
the gain resistors will have no effect when amplifying the 6.25x volt

zener

up to 10.000000 Volts.
(or times 10/7 for use with other types of voltage references)

ws


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Zbyszek


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Zbyszek

I would e grateful for a copy of such documentation. By the way; Weston saturated standard cells are still in use in our polish equivalent of american NBS; they kept about 400 units in Glowny Urzad Miar in Warsaw. In 1937 modified Weston cell was build it has 11% Cd 15% Bi and remain Hg as amalgamate. It had 13 ppm tempo almoust 5 time less than original Weston cell. In 1964 Roger G. Bates wrote a paper regarding stability o a such secondary standard; the EFM of such cell was 1.0188 V at 25 Cel.deg., Six cells after 26 years changed their EMF average by 0.03 mV. Because I am biochemist, so electrochemisty was slightly close to my interest. Sometimes I am dreaming to build such a cell... Best regards, from Warsaw, 2011/9/16 Fred Schneider <pa4tim@gmail.com> > I have the original guildline enclosure with schematic. There is an oven in > there and the circuit is rather straight forward using some tranistors. > Maybe you can use this as an example. The saturated cells must be hold at 30 > degrees. They must kept vertical allways and viberation free. After moving > them arround they need about two weeks to stabilize and do not use a 10M > meter on them. Mine are from 1971 and still in very good shape. > > Fred PA4TIM > > Op 16 sep. 2011 om 00:03 heeft zbigniew169 <zbigniew169@gmail.com> het > volgende geschreven: > > > He I am Zbigniew W. Kaminski from Warsaw Poland. I was member-observer up > to > > this time. I am user of Prema 6040D 7 & 1/2 digits multimeter. > > > > I would be appreciate if You will send some details concerning > construction > > of Your mini-oven. I need something similar for Weston saturated standard > > cell. In my case the stability must be even better than 0.1 Celsjus > degree, > > because of Weston temco in the range about 50 ppm. I think that in my > > project I should use patinium resistor rathen than termistor, because > it's > > much long term stable than termistor. Platinium element is much less > > sensitive to temperature, therefore the cuircuit will be much > complicated. > > I am an owner of single Weston cell which stays in my lab during last 20 > > years, and I am curious how stable it is. > > > > 2011/9/15 WarrenS <warrensjmail-one@yahoo.com> > > > >> > >> It is well known that by changing the nominal operating current of a > 1N825 > >> type reference diode, that its voltage turn over temperature point can > be > >> moved. > >> Usually the turn over temperature can be set to the nominal operating > >> temperature, there by giving a Zero TC reference voltage over a limited > >> temperature range. > >> For the parts I'm testing, the voltage changes less than 0.1 PPM with a > 5 > >> deg F temperature variation, and 1PPM over a 20 deg F range when the > zero TC > >> current is set for room temperature. > >> Not bad, but being an extreme Volt Nut, I always want to do better. > >> > >> Two ways to make it better, Add an X^2 (second order) temperature > >> compensator or put the reference in an oven. > >> > >> What I did was make a very simple Mini-Oven using an 1N825, a glass > >> thermister & a 1/4 resistor, > >> all heat shrink together in a small package about the size of a 1 watt > >> resistor. > >> Now by Appling about 1/8 W nominal power to the resistor from a simple > >> temperature controller, the thermistor can be held at a constant > >> Temperature. > >> This allows setting it up so that normal temperature changes have true > ZERO > >> effect on the zero TCed zener voltage. > >> Next thing I want to make is a simple times 8/5 digital gain amp so that > >> the gain resistors will have no effect when amplifying the 6.25x volt > zener > >> up to 10.000000 Volts. > >> (or times 10/7 for use with other types of voltage references) > >> > >> ws > >> > >> _______________________________________________ > >> volt-nuts mailing list -- volt-nuts@febo.com > >> To unsubscribe, go to > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > >> and follow the instructions there. > >> > > > > > > > > -- > > Zbyszek > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > -- Zbyszek
FS
Fred Schneider
Fri, Sep 16, 2011 5:56 AM

Ok, i will look it up for you and mail you,  it is on my other computer.
http://www.pa4tim.nl/?p=2141 these are my cells. ( text is dutch but there is a Translate button on the right menubar)
My best cells changed about the same rate as you wrote. I have all calibrations from new upto theire retirement in 1988. Then the lab from that company was closed but it stayed there,  unpowered, on the same shelf until this year. The lab was , over he years, turned in to a climatized computerroom. Now they needed space so the still remainig calibration stuff ( including a 19 inch rack with Flukes) had to be destroyed.  A friend who workes there in a high function could not let that happen and did a lot of work to try and save as much as possible. Still a lot is distroyed, like precision resistor decades, the Fluke line compensator and lots of spare parts, but what is saved he brought to me. Besides the Guildline the Fluke 720, 750, 760, 731A, 535, 332, 8500 some HP stuff like a Hi- Z source meter, decade capacitor and sinewave generator and a Tek 576 curvetracer.

Fred PA4TIM

Op 16 sep. 2011 om 01:21 heeft zbigniew169 zbigniew169@gmail.com het volgende geschreven:

I would e grateful for a copy of such documentation.

By the way; Weston saturated standard cells are still in use in our polish
equivalent of american NBS; they kept about 400 units in Glowny Urzad Miar
in Warsaw.
In 1937 modified Weston cell was build it has 11% Cd 15% Bi and remain Hg as
amalgamate. It had 13 ppm tempo almoust 5 time less than original Weston
cell. In 1964 Roger G. Bates wrote a paper regarding stability o a such
secondary standard; the EFM of such cell was 1.0188 V at 25 Cel.deg., Six
cells after 26 years changed their EMF average by 0.03 mV. Because I am
biochemist, so electrochemisty was slightly close to my interest. Sometimes
I am dreaming to build such a cell...

Best regards, from Warsaw,

2011/9/16 Fred Schneider pa4tim@gmail.com

I have the original guildline enclosure with schematic. There is an oven in
there and the circuit is rather straight forward using some tranistors.
Maybe you can use this as an example. The saturated cells must be hold at 30
degrees. They must kept vertical allways and viberation free. After moving
them arround they need about two weeks to stabilize and do not use a 10M
meter on them. Mine are from 1971 and still in very good shape.

Fred PA4TIM

Op 16 sep. 2011 om 00:03 heeft zbigniew169 zbigniew169@gmail.com het
volgende geschreven:

He I am Zbigniew W. Kaminski from Warsaw Poland. I was member-observer up

to

this time. I am user of Prema 6040D 7 & 1/2 digits multimeter.

I would be appreciate if You will send some details concerning

construction

of Your mini-oven. I need something similar for Weston saturated standard
cell. In my case the stability must be even better than 0.1 Celsjus

degree,

because of Weston temco in the range about 50 ppm. I think that in my
project I should use patinium resistor rathen than termistor, because

it's

much long term stable than termistor. Platinium element is much less
sensitive to temperature, therefore the cuircuit  will be much

complicated.

I am an owner of single Weston cell which stays in my lab during last 20
years, and I am curious how stable it is.

2011/9/15 WarrenS warrensjmail-one@yahoo.com

It is well known that by changing the nominal operating current of a

1N825

type reference diode, that its voltage turn over temperature point can

be

moved.
Usually the turn over temperature can be set to the nominal operating
temperature,  there by giving a Zero TC reference voltage over a limited
temperature range.
For the parts I'm testing, the voltage changes less than 0.1 PPM with a

5

deg F temperature variation, and 1PPM over a 20 deg F range when the

zero TC

current is set for room temperature.
Not bad, but being an extreme Volt Nut, I always want to do better.

Two ways to make it better, Add an X^2 (second order) temperature
compensator or put the reference in an oven.

What I did was make a very simple Mini-Oven using an 1N825, a glass
thermister & a 1/4 resistor,
all heat shrink together  in a small package about the size of a 1 watt
resistor.
Now by Appling about 1/8 W nominal power to the resistor from a simple
temperature controller, the thermistor can be held at a constant
Temperature.
This allows setting it up so that normal temperature changes have true

ZERO

effect on the zero TCed zener voltage.
Next thing I want to make is a simple times 8/5 digital gain amp so that
the gain resistors will have no effect when amplifying the 6.25x volt

zener

up to 10.000000 Volts.
(or times 10/7 for use with other types of voltage references)

ws


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Zbyszek


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

--
Zbyszek


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Ok, i will look it up for you and mail you, it is on my other computer. http://www.pa4tim.nl/?p=2141 these are my cells. ( text is dutch but there is a Translate button on the right menubar) My best cells changed about the same rate as you wrote. I have all calibrations from new upto theire retirement in 1988. Then the lab from that company was closed but it stayed there, unpowered, on the same shelf until this year. The lab was , over he years, turned in to a climatized computerroom. Now they needed space so the still remainig calibration stuff ( including a 19 inch rack with Flukes) had to be destroyed. A friend who workes there in a high function could not let that happen and did a lot of work to try and save as much as possible. Still a lot is distroyed, like precision resistor decades, the Fluke line compensator and lots of spare parts, but what is saved he brought to me. Besides the Guildline the Fluke 720, 750, 760, 731A, 535, 332, 8500 some HP stuff like a Hi- Z source meter, decade capacitor and sinewave generator and a Tek 576 curvetracer. Fred PA4TIM Op 16 sep. 2011 om 01:21 heeft zbigniew169 <zbigniew169@gmail.com> het volgende geschreven: > I would e grateful for a copy of such documentation. > > By the way; Weston saturated standard cells are still in use in our polish > equivalent of american NBS; they kept about 400 units in Glowny Urzad Miar > in Warsaw. > In 1937 modified Weston cell was build it has 11% Cd 15% Bi and remain Hg as > amalgamate. It had 13 ppm tempo almoust 5 time less than original Weston > cell. In 1964 Roger G. Bates wrote a paper regarding stability o a such > secondary standard; the EFM of such cell was 1.0188 V at 25 Cel.deg., Six > cells after 26 years changed their EMF average by 0.03 mV. Because I am > biochemist, so electrochemisty was slightly close to my interest. Sometimes > I am dreaming to build such a cell... > > Best regards, from Warsaw, > > 2011/9/16 Fred Schneider <pa4tim@gmail.com> > >> I have the original guildline enclosure with schematic. There is an oven in >> there and the circuit is rather straight forward using some tranistors. >> Maybe you can use this as an example. The saturated cells must be hold at 30 >> degrees. They must kept vertical allways and viberation free. After moving >> them arround they need about two weeks to stabilize and do not use a 10M >> meter on them. Mine are from 1971 and still in very good shape. >> >> Fred PA4TIM >> >> Op 16 sep. 2011 om 00:03 heeft zbigniew169 <zbigniew169@gmail.com> het >> volgende geschreven: >> >>> He I am Zbigniew W. Kaminski from Warsaw Poland. I was member-observer up >> to >>> this time. I am user of Prema 6040D 7 & 1/2 digits multimeter. >>> >>> I would be appreciate if You will send some details concerning >> construction >>> of Your mini-oven. I need something similar for Weston saturated standard >>> cell. In my case the stability must be even better than 0.1 Celsjus >> degree, >>> because of Weston temco in the range about 50 ppm. I think that in my >>> project I should use patinium resistor rathen than termistor, because >> it's >>> much long term stable than termistor. Platinium element is much less >>> sensitive to temperature, therefore the cuircuit will be much >> complicated. >>> I am an owner of single Weston cell which stays in my lab during last 20 >>> years, and I am curious how stable it is. >>> >>> 2011/9/15 WarrenS <warrensjmail-one@yahoo.com> >>> >>>> >>>> It is well known that by changing the nominal operating current of a >> 1N825 >>>> type reference diode, that its voltage turn over temperature point can >> be >>>> moved. >>>> Usually the turn over temperature can be set to the nominal operating >>>> temperature, there by giving a Zero TC reference voltage over a limited >>>> temperature range. >>>> For the parts I'm testing, the voltage changes less than 0.1 PPM with a >> 5 >>>> deg F temperature variation, and 1PPM over a 20 deg F range when the >> zero TC >>>> current is set for room temperature. >>>> Not bad, but being an extreme Volt Nut, I always want to do better. >>>> >>>> Two ways to make it better, Add an X^2 (second order) temperature >>>> compensator or put the reference in an oven. >>>> >>>> What I did was make a very simple Mini-Oven using an 1N825, a glass >>>> thermister & a 1/4 resistor, >>>> all heat shrink together in a small package about the size of a 1 watt >>>> resistor. >>>> Now by Appling about 1/8 W nominal power to the resistor from a simple >>>> temperature controller, the thermistor can be held at a constant >>>> Temperature. >>>> This allows setting it up so that normal temperature changes have true >> ZERO >>>> effect on the zero TCed zener voltage. >>>> Next thing I want to make is a simple times 8/5 digital gain amp so that >>>> the gain resistors will have no effect when amplifying the 6.25x volt >> zener >>>> up to 10.000000 Volts. >>>> (or times 10/7 for use with other types of voltage references) >>>> >>>> ws >>>> >>>> _______________________________________________ >>>> volt-nuts mailing list -- volt-nuts@febo.com >>>> To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >>>> and follow the instructions there. >>>> >>> >>> >>> >>> -- >>> Zbyszek >>> _______________________________________________ >>> volt-nuts mailing list -- volt-nuts@febo.com >>> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> > > > > -- > Zbyszek > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.