I recently repaired a Weston 4442 (3.5 digit dual-slope DVM) which had
a failure of the integrating cap. It had decreased in value from 2.2
uf to 0.42 uf; a radial-lead green "box" containing an unknown
dielectric) Not having a suitable polystyrene or Teflon cap in the
junk box, I used a handy polyester cap (looks like an orange-drop but
white epoxy).
Any ideas as to how much error the (presumably) Mylar integrator cap
will introduce? In other words, will such error be significant in a
mere 3.5 digits? (I imagine that Weston would have used a less
expensive cap if it would do the job). The absolute value isn't
important due to the dual-slope technique, but time-variant storage
effects are...
I have an HP 4.5 digit DVM and a MAX6385 (2.5 volt) reference, but no
"volt-nuts"-league meters or standards, other than a 596.5 ohm 0.01%
resistor ;)
thanks
Charles
Charles Morris wrote:
I recently repaired a Weston 4442 (3.5 digit dual-slope DVM) which had
a failure of the integrating cap. It had decreased in value from 2.2
uf to 0.42 uf; a radial-lead green "box" containing an unknown
dielectric) Not having a suitable polystyrene or Teflon cap in the
junk box, I used a handy polyester cap (looks like an orange-drop but
white epoxy).
Any ideas as to how much error the (presumably) Mylar integrator cap
will introduce? In other words, will such error be significant in a
mere 3.5 digits? (I imagine that Weston would have used a less
expensive cap if it would do the job). The absolute value isn't
important due to the dual-slope technique, but time-variant storage
effects are...
I have an HP 4.5 digit DVM and a MAX6385 (2.5 volt) reference, but no
"volt-nuts"-league meters or standards, other than a 596.5 ohm 0.01%
resistor ;)
thanks
Charles
The effect will depend on the dual slope integration time and could be
as high as 1% depending on the particular capacitor.
Bruce
Any ideas as to how much error the (presumably) Mylar integrator cap
will introduce? In other words, will such error be significant in a
mere 3.5 digits?
Hello,
several years ago I built a integrating ADC.
2.5V Reference as "zero voltage" and integrating +5V and 0V
against the input signal.
(0 .. 5V measured as -2.5 to 2.5 V).
Due to capacitor soaking I had a gap in the middle of the 0 to 5V
range which was about 1mV (0.02%) as non linear jump.
I could reduce the non linearity by replacing the mylar capacitor with
a polypropylene capacitor by a factor of around 10.
So for a meter with 3,5 digits this should not play a role.
With best regards
Andreas