Discussion of precise voltage measurement

View all threads

Finally got around to modifying my Fluke 845ab with LED's

TH
Tony Holt
Thu, Sep 11, 2014 8:14 PM

On 11/09/2014 01:50, Mike S wrote:

On 9/10/2014 7:00 PM, Tony wrote:

I've just noticed that TI and Linear's specs for 'Long Term Stability'
(typical) are different. TI state 20ppm/1000Hr while Linear state
8ppm/SQRT(kHr). That's  a big difference - is this likely to be a real
difference or just specmanship?

I note that Linear (in Note 4) also state that "Devices with maximum
guaranteed long-term stability of 20ppm/SQRT(kH) are available."
Presumably they would be a special order as there doesn't appear to be a
unique part no. Would they be likely to be much more expensive?

Isn't 8ppm/SQRT(kHr) better than 20ppm/SQRT(kH)? Why would the latter
be more expensive? Or is it the difference between "typical" and
"guaranteed?"

I'm guessing that typical in this case means the one sigma value so the
three sigma value would be 24ppm. In any case three sigma still only
means 93.32% of parts come within that limit, or 6.7% exceed 24ppm, and
a few could be considerably worse. That compares to a guarantee that all
all parts meet 20ppm. This link:
http://www.gellerlabs.com/LM299AH-20_Case_Study.htm provided in
Andreas's response is very interesting:

"Certified Long Term Drift The National Semiconductor LM199AH-20,
LM299AH-20, and LM399AH-50 are ultra-stable Zener references specially
selected from the production runs of LM199AH, LM299AH, LM399AH and
tested to confirm a long-term stability of 20, 20, or 50 ppm per 1000
hours, respectively..."

So in this case they really do mean a guarantee. And I doubt that
individual testing came cheap. I say 'came' because I wonder if they
still 100% test the 20ppm parts or if they select them using some lower
costs means?

Tony H

On 11/09/2014 01:50, Mike S wrote: > On 9/10/2014 7:00 PM, Tony wrote: >> I've just noticed that TI and Linear's specs for 'Long Term Stability' >> (typical) are different. TI state 20ppm/1000Hr while Linear state >> 8ppm/SQRT(kHr). That's a big difference - is this likely to be a real >> difference or just specmanship? >> >> I note that Linear (in Note 4) also state that "Devices with maximum >> guaranteed long-term stability of 20ppm/SQRT(kH) are available." >> Presumably they would be a special order as there doesn't appear to be a >> unique part no. Would they be likely to be much more expensive? > > Isn't 8ppm/SQRT(kHr) better than 20ppm/SQRT(kH)? Why would the latter > be more expensive? Or is it the difference between "typical" and > "guaranteed?" > I'm guessing that typical in this case means the one sigma value so the three sigma value would be 24ppm. In any case three sigma still only means 93.32% of parts come within that limit, or 6.7% exceed 24ppm, and a few could be considerably worse. That compares to a guarantee that all all parts meet 20ppm. This link: http://www.gellerlabs.com/LM299AH-20_Case_Study.htm provided in Andreas's response is very interesting: "Certified Long Term Drift The National Semiconductor LM199AH-20, LM299AH-20, and LM399AH-50 are ultra-stable Zener references specially selected from the production runs of LM199AH, LM299AH, LM399AH and tested to confirm a long-term stability of 20, 20, or 50 ppm per 1000 hours, respectively..." So in this case they really do mean a guarantee. And I doubt that individual testing came cheap. I say 'came' because I wonder if they still 100% test the 20ppm parts or if they select them using some lower costs means? Tony H
TH
Tony Holt
Thu, Sep 11, 2014 11:06 PM

Andreas,

Thanks for taking the time to respond. Actually I've seen many of your
postings on eevblog and here - you've clearly done a great deal of work
in this area and would like to thank you for making it available to us all.

On 11/09/2014 06:07, Andreas Jahn wrote:

Hello,

many questions I will keep it short:

All ageing specs are "typical" if you want to have "guaranteed" values
you will have to measure it over a reasonable time. (I recommend min 6
months).
Every treatment (soldering, mechanical/temperature shock) of a
reference may create a new ageing cycle with different slope.

True. I guess that the new ageing cycle from soldering in an LM399 is
not going to be as bad as that for a surface mount plastic device.

So 100ppm/15 years outside of "lab conditions" (23 deg , constant
humidity) is something that I would not guarantee without re-calibration.

I had a feeling that would be the answer - though surely humidity
shouldn't be a factor as these are hermetic parts. The questions remains
though, what level might you specify - if you were forced to come up
with a number (ok a guess!) - for non-selected, non-pre-aged parts after
15years continuous operation without re-calibration? Obviously this is
given the context of the presumably limited numbers of samples you've
tested and I guess you wouldn't have bothered to further test early
rejects.

Although typical drift of pre-aged + selected references will be in
the 1-2ppm/year range if properly treated.

What would you classify as pre-aged? Do they need to be powered up or
can they be maintained at a suitable temperature? How many rejects would
you expect to get to get one that achieves 1-2ppm?

Is it known if the major instrument manufacturers preselect and burn-in
LM399s themselves for their middle-range instruments? I'm pretty sure
the top end kit will be all use carefully tested and selected parts, but
what about a 34401A for example? The basic accuracy spec for that is
20ppm for 90 days, 35ppm for 12 months so even a 20ppm guaranteed part
wouldn't be good enough, especially allowing margin for drift in other
components. I guess I just answered my own question!

Also its meaningless if you want to have LT or National (TI) parts
since LT is the only manufacturer which still produces them.
With high demands you will also have to sort out the "noisy" references.

Some "typical" LM399 (all from NS) ageing data can be found on web:

http://www.gellerlabs.com/LM299AH-20_Case_Study.htm

That's very interesting. I have to agree that the raw data looks
suspect. I wonder what the rejection rate is for this 20ppm selection
and does it mean that non-selected parts have a high probability of
being worse than 20ppm?

I just came across another part which looks very interesting given its
low cost - the automotive qualified REF5050-Q1. Although its only spec'd
as 3ppm/C typical, 8ppm/C max, that's using the box method over -40 to
+125C. The 'typical' chart however, figure 4, page 5 shows the gradients
to be very flat between 25 and 50. Its typical of course, so real parts
may be very different aka Vishay foil resistors. The 0 to 85C histogram,
fig 1 on page 5, do show the majority of parts being in the range
.75ppm/C to 1.75ppm/C which is pretty good, and with luck, in the 25 to
50C range may well be much better so a crude heating arrangement may be
worthwhile (made easier by the 5050's temperature output!)

I can't reconcile fig 4 with the histograms though; from the chart I
reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40 to
125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show modal
values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong or are
these specs inconsistent?

Even more surprising is the headline feature on page 1:

"EXCELLENT LONG-TERM STABILITY:
– 5 ppm/1000 hr (typ) after 1000 hours"

Unfortunately that seems to be an error as the 'typical' spec on page 4 is:
90ppm (0-1000 hours)
10ppm (1000 to 2000 hours).

The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96 parts
show the worst case being +25ppm, with the bulk ending approx between 0
and 15ppm. I wonder if they carry on improving after 2kHrs?

That's definitely not the SQRT(1kHr) characteristic and is very
different from the standard REF5050 which quotes 100ppm (1st 100hours),
50ppm (1000 to 2000 hours).

If you are in a position to pre-age them for 1000 hours that 10ppm spec
is almost as good as the LM399 and best of all, TI quote a price of
$1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from Linear. One off
prices are rather more at $4.15 from Digikey (part no REF5050AQDRQ1) but
again is still a lot cheaper than an LM399 at $9.92. At $1.60 and .8mA
supply current, using 4, 8 or even dozens is a realistic proposition to
exploit statistical improvements and noise reduction.

Noise is a bit high at 15uVpp. They're also trimmable. Shame there isn't
an hermetic part though.

Anybody tried these or spotted the gotchas? Alternatively has anyone
here evaluated the hermetic LTC6655 for long term drift?

Tony H

Am 11.09.2014 um 01:00 schrieb Tony:

I've just noticed that TI and Linear's specs for 'Long Term
Stability' (typical) are different. TI state 20ppm/1000Hr while
Linear state 8ppm/SQRT(kHr). That's  a big difference - is this
likely to be a real difference or just specmanship?

I note that Linear (in Note 4) also state that "Devices with maximum
guaranteed long-term stability of 20ppm/SQRT(kH) are available."
Presumably they would be a special order as there doesn't appear to
be a unique part no. Would they be likely to be much more expensive?

Then on page 4 Linear show a graph of long term performance of 44
units (rather cheekily starting the graph at 2 months or approx 1500
hours!). To reproduce something approaching the mean curve using the
formulae (drift ppm/SQRT(kHr)) x SQRT(month * 24 * 30.5/1000),
requires me to use 2.2ppm/SQRT(kHr). That is way less than the
typical 8.5ppm value.

To get a curve that resembles the 3-sigma curve requires a value of
5.7ppm/SQRT(kHr) which is still better than that 8ppm typical figure.

I'm not sure how to interpret this; what value would you use if you
were designing a reference that isn't going to be re-calibrated after
the initial calibration and you don't intend to burn in for several
months?

Assuming the equipment is expected to have a 15 year life, operating
in a range of 0 to 40C, what maximum total drift would you be
comfortable specifying? I'd prefer it to be less than 100ppm, but
that would require a drift of < 9ppm/SQRT(kHr), but that assumes that
the SQRT(KHr) drift characteristic is valid for periods much longer
than 12 months.

Are there any other references, at similiar or lower cost, that could
be reasonably guaranteed to have a total drift of < 100ppm after
15years?

Is it reasonable to assume that there are some types of voltage
reference will always drift, albeit noisily, in one direction
allowing the original calibration to be offset to some extent to
reduce the maximum error over its lifetime?

Having looked at several application notes and lots of datasheets, in
those that include graphs of drift over 1kHrs or so of several
'typical' examples, I have not been able to see any meaningful
correlation between the specified typical 1k drift figures and the
graphs. Eg. in Linear's Design Note 229 (Don't Be Fooled By Voltage
Reference Long-Term Drift and Hysteresis" the graphs of drift for the
LT1461S8 and the LT1790SOT23 show very different drift after 1600
hours - in the range 50 to 130 for the former and approx -5 to +45
for the latter, yet the LT1461 is spec'd at 60ppm/SQRT(1kHr) and the
LT17910 at 50ppm.

I realise that I would probably need to contact the manufacturers for
real answers but its been my experience that they aren't often
interested if you're not buying large volumes, and I know that a lot
of people here have a lot of experience in this area.

Thanks,
Tony H


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Andreas, Thanks for taking the time to respond. Actually I've seen many of your postings on eevblog and here - you've clearly done a great deal of work in this area and would like to thank you for making it available to us all. On 11/09/2014 06:07, Andreas Jahn wrote: > Hello, > > many questions I will keep it short: > > All ageing specs are "typical" if you want to have "guaranteed" values > you will have to measure it over a reasonable time. (I recommend min 6 > months). > Every treatment (soldering, mechanical/temperature shock) of a > reference may create a new ageing cycle with different slope. True. I guess that the new ageing cycle from soldering in an LM399 is not going to be as bad as that for a surface mount plastic device. > > So 100ppm/15 years outside of "lab conditions" (23 deg , constant > humidity) is something that I would not guarantee without re-calibration. I had a feeling that would be the answer - though surely humidity shouldn't be a factor as these are hermetic parts. The questions remains though, what level might you specify - if you were forced to come up with a number (ok a guess!) - for non-selected, non-pre-aged parts after 15years continuous operation without re-calibration? Obviously this is given the context of the presumably limited numbers of samples you've tested and I guess you wouldn't have bothered to further test early rejects. > Although typical drift of pre-aged + selected references will be in > the 1-2ppm/year range if properly treated. What would you classify as pre-aged? Do they need to be powered up or can they be maintained at a suitable temperature? How many rejects would you expect to get to get one that achieves 1-2ppm? Is it known if the major instrument manufacturers preselect and burn-in LM399s themselves for their middle-range instruments? I'm pretty sure the top end kit will be all use carefully tested and selected parts, but what about a 34401A for example? The basic accuracy spec for that is 20ppm for 90 days, 35ppm for 12 months so even a 20ppm guaranteed part wouldn't be good enough, especially allowing margin for drift in other components. I guess I just answered my own question! > > Also its meaningless if you want to have LT or National (TI) parts > since LT is the only manufacturer which still produces them. > With high demands you will also have to sort out the "noisy" references. > > Some "typical" LM399 (all from NS) ageing data can be found on web: > > http://www.gellerlabs.com/LM299AH-20_Case_Study.htm That's very interesting. I have to agree that the raw data looks suspect. I wonder what the rejection rate is for this 20ppm selection and does it mean that non-selected parts have a high probability of being worse than 20ppm? > http://www.eevblog.com/forum/projects/lm399-based-10-v-reference/msg478496/#msg478496 > > > With best regards > > Andreas > I just came across another part which looks very interesting given its low cost - the automotive qualified REF5050-Q1. Although its only spec'd as 3ppm/C typical, 8ppm/C max, that's using the box method over -40 to +125C. The 'typical' chart however, figure 4, page 5 shows the gradients to be very flat between 25 and 50. Its typical of course, so real parts may be very different aka Vishay foil resistors. The 0 to 85C histogram, fig 1 on page 5, do show the majority of parts being in the range .75ppm/C to 1.75ppm/C which is pretty good, and with luck, in the 25 to 50C range may well be much better so a crude heating arrangement may be worthwhile (made easier by the 5050's temperature output!) I can't reconcile fig 4 with the histograms though; from the chart I reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40 to 125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show modal values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong or are these specs inconsistent? Even more surprising is the headline feature on page 1: "EXCELLENT LONG-TERM STABILITY: – 5 ppm/1000 hr (typ) after 1000 hours" Unfortunately that seems to be an error as the 'typical' spec on page 4 is: 90ppm (0-1000 hours) 10ppm (1000 to 2000 hours). The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96 parts show the worst case being +25ppm, with the bulk ending approx between 0 and 15ppm. I wonder if they carry on improving after 2kHrs? That's definitely not the SQRT(1kHr) characteristic and is very different from the standard REF5050 which quotes 100ppm (1st 100hours), 50ppm (1000 to 2000 hours). If you are in a position to pre-age them for 1000 hours that 10ppm spec is almost as good as the LM399 and best of all, TI quote a price of $1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from Linear. One off prices are rather more at $4.15 from Digikey (part no REF5050AQDRQ1) but again is still a lot cheaper than an LM399 at $9.92. At $1.60 and .8mA supply current, using 4, 8 or even dozens is a realistic proposition to exploit statistical improvements and noise reduction. Noise is a bit high at 15uVpp. They're also trimmable. Shame there isn't an hermetic part though. Anybody tried these or spotted the gotchas? Alternatively has anyone here evaluated the hermetic LTC6655 for long term drift? Tony H > Am 11.09.2014 um 01:00 schrieb Tony: >> I've just noticed that TI and Linear's specs for 'Long Term >> Stability' (typical) are different. TI state 20ppm/1000Hr while >> Linear state 8ppm/SQRT(kHr). That's a big difference - is this >> likely to be a real difference or just specmanship? >> >> I note that Linear (in Note 4) also state that "Devices with maximum >> guaranteed long-term stability of 20ppm/SQRT(kH) are available." >> Presumably they would be a special order as there doesn't appear to >> be a unique part no. Would they be likely to be much more expensive? >> >> Then on page 4 Linear show a graph of long term performance of 44 >> units (rather cheekily starting the graph at 2 months or approx 1500 >> hours!). To reproduce something approaching the mean curve using the >> formulae (drift ppm/SQRT(kHr)) x SQRT(month * 24 * 30.5/1000), >> requires me to use 2.2ppm/SQRT(kHr). That is way less than the >> typical 8.5ppm value. >> >> To get a curve that resembles the 3-sigma curve requires a value of >> 5.7ppm/SQRT(kHr) which is still better than that 8ppm typical figure. >> >> I'm not sure how to interpret this; what value would you use if you >> were designing a reference that isn't going to be re-calibrated after >> the initial calibration and you don't intend to burn in for several >> months? >> >> Assuming the equipment is expected to have a 15 year life, operating >> in a range of 0 to 40C, what maximum total drift would you be >> comfortable specifying? I'd prefer it to be less than 100ppm, but >> that would require a drift of < 9ppm/SQRT(kHr), but that assumes that >> the SQRT(KHr) drift characteristic is valid for periods much longer >> than 12 months. >> >> Are there any other references, at similiar or lower cost, that could >> be reasonably guaranteed to have a total drift of < 100ppm after >> 15years? >> >> Is it reasonable to assume that there are some types of voltage >> reference will always drift, albeit noisily, in one direction >> allowing the original calibration to be offset to some extent to >> reduce the maximum error over its lifetime? >> >> Having looked at several application notes and lots of datasheets, in >> those that include graphs of drift over 1kHrs or so of several >> 'typical' examples, I have not been able to see any meaningful >> correlation between the specified typical 1k drift figures and the >> graphs. Eg. in Linear's Design Note 229 (Don't Be Fooled By Voltage >> Reference Long-Term Drift and Hysteresis" the graphs of drift for the >> LT1461S8 and the LT1790SOT23 show very different drift after 1600 >> hours - in the range 50 to 130 for the former and approx -5 to +45 >> for the latter, yet the LT1461 is spec'd at 60ppm/SQRT(1kHr) and the >> LT17910 at 50ppm. >> >> I realise that I would probably need to contact the manufacturers for >> real answers but its been my experience that they aren't often >> interested if you're not buying large volumes, and I know that a lot >> of people here have a lot of experience in this area. >> >> Thanks, >> Tony H >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
AJ
Andreas Jahn
Fri, Sep 12, 2014 5:16 AM

Am 12.09.2014 um 01:06 schrieb Tony Holt:

All ageing specs are "typical" if you want to have "guaranteed"
values you will have to measure it over a reasonable time. (I
recommend min 6 months).
Every treatment (soldering, mechanical/temperature shock) of a
reference may create a new ageing cycle with different slope.

True. I guess that the new ageing cycle from soldering in an LM399 is
not going to be as bad as that for a surface mount plastic device.

Thats true humidity within plastic housing is changed largely by
soldering. This gives more stress on the die than on hermetically devices.
But why are the instrument manufacturers using sockets for the LM399
(HP34401A) or reference boards for the LTZ1000 for their pre selected parts?

So 100ppm/15 years outside of "lab conditions" (23 deg , constant
humidity) is something that I would not guarantee without
re-calibration.

I had a feeling that would be the answer - though surely humidity
shouldn't be a factor as these are hermetic parts. The questions
remains though, what level might you specify - if you were forced to
come up with a number (ok a guess!) - for non-selected, non-pre-aged
parts after 15years continuous operation without re-calibration?
Obviously this is given the context of the presumably limited numbers
of samples you've tested and I guess you wouldn't have bothered to
further test early rejects.

Sorry this may change from lot to lot. From other (non heated)
references I have very different results. Especially from devices bought
before and after ROHS. And partly even better graded devices (LT1027BCN)
behave worse than standard grade devices (LT1027CCN) in my temperature
range.
With humidity and LM399: this should be no issue for the metal can
package of the LM399. Although I have one LM399 (#1) which I desoldered
from a board of unknown age that has a correlation to humidity (see
attachment). Ageing is also in the range of 12-15 ppm/year average with
seasonal modulation which follows humidity. But since this behavior is
not typical for my other LM399 references I would justify this sample as
"defect".

Humidity: even hermetically parts can suffer from humidity. The epoxy
board swells if exposed to humidity. This creates stress to the leads of
the package and then to the die. I measure around 0.5ppm/% for plastic
parts. In a early publication LT specced around 12ppm change for a 30%
rH change for the LT1236AILS8 hermetically package. This spec is now
removed from newer data sheets. You will only find a hint to avoid board
stress.

Although typical drift of pre-aged + selected references will be in
the 1-2ppm/year range if properly treated.

What would you classify as pre-aged? Do they need to be powered up or
can they be maintained at a suitable temperature? How many rejects
would you expect to get to get one that achieves 1-2ppm?

Is it known if the major instrument manufacturers preselect and
burn-in LM399s themselves for their middle-range instruments? I'm
pretty sure the top end kit will be all use carefully tested and
selected parts, but what about a 34401A for example? The basic
accuracy spec for that is 20ppm for 90 days, 35ppm for 12 months so
even a 20ppm guaranteed part wouldn't be good enough, especially
allowing margin for drift in other components. I guess I just answered
my own question!

From Keithley cal lab you can see that they adjust the instrument
during calibration if they fall outside the 70% window. So I guess that
after 1 year they think that ageing of the components (including the
reference) is below 30% of the 35 ppm spec per year. The part numbers of
the LM399 are not the original ones but a own manufacturer specific.
So they do a pre-ageing. Of course powered up. (perhaps intermittend
like on resistors?)

I just came across another part which looks very interesting given its
low cost - the automotive qualified REF5050-Q1. Although its only
spec'd as 3ppm/C typical, 8ppm/C max, that's using the box method over
-40 to +125C. The 'typical' chart however, figure 4, page 5 shows the
gradients to be very flat between 25 and 50. Its typical of course, so
real parts may be very different aka Vishay foil resistors. The 0 to
85C histogram, fig 1 on page 5, do show the majority of parts being in
the range .75ppm/C to 1.75ppm/C which is pretty good, and with luck,
in the 25 to 50C range may well be much better so a crude heating
arrangement may be worthwhile (made easier by the 5050's temperature
output!)

I can't reconcile fig 4 with the histograms though; from the chart I
reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40 to
125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show modal
values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong or are
these specs inconsistent?

Even more surprising is the headline feature on page 1:

"EXCELLENT LONG-TERM STABILITY:
– 5 ppm/1000 hr (typ) after 1000 hours"

Unfortunately that seems to be an error as the 'typical' spec on page
4 is:
90ppm (0-1000 hours)
10ppm (1000 to 2000 hours).

The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96 parts
show the worst case being +25ppm, with the bulk ending approx between
0 and 15ppm. I wonder if they carry on improving after 2kHrs?

That's definitely not the SQRT(1kHr) characteristic and is very
different from the standard REF5050 which quotes 100ppm (1st
100hours), 50ppm (1000 to 2000 hours).

If you are in a position to pre-age them for 1000 hours that 10ppm
spec is almost as good as the LM399 and best of all, TI quote a price
of $1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from Linear.
One off prices are rather more at $4.15 from Digikey (part no
REF5050AQDRQ1) but again is still a lot cheaper than an LM399 at
$9.92. At $1.60 and .8mA supply current, using 4, 8 or even dozens is
a realistic proposition to exploit statistical improvements and noise
reduction.

Noise is a bit high at 15uVpp. They're also trimmable. Shame there
isn't an hermetic part though.

Anybody tried these or spotted the gotchas? Alternatively has anyone
here evaluated the hermetic LTC6655 for long term drift?

Tony H

Do you really consider unheated references for a 100 ppm lifetime spec?
And surface mount SO-8 devices which have a factor 3 worse data on
hysteresis (and probably humidity) than the DIP-8 package?
I do not know where they get their "typical" T.C. curves with flat area
near 25 degrees. Usually they are only measuring 3 points of the
references. (-40, 25, +125 deg C or whatever the temperature range of
the device is).
From theory any bandgap device (REF5050, LTC6655) has to have worse
ageing than a buried zener (LM399, AD586). But also the manufacturer
specific packaging has its influence. The specs in the datasheets get
better the later the datasheets are created for several reasons: The
instruments nowadays are with less tolerances. The silicon for the dies
is more pure than the elder qualified devices. And of course then
marketing demands that every generation has to be better than the
previous one. But this has nothing to do with physics and the actual
behaviour of the devices.
I have my first sample of LTC6655AILS8 measured for T.C. this week. But
for my needs it has a too large hysteresis.

With best regards

Andreas

Am 12.09.2014 um 01:06 schrieb Tony Holt: > >> All ageing specs are "typical" if you want to have "guaranteed" >> values you will have to measure it over a reasonable time. (I >> recommend min 6 months). >> Every treatment (soldering, mechanical/temperature shock) of a >> reference may create a new ageing cycle with different slope. > > True. I guess that the new ageing cycle from soldering in an LM399 is > not going to be as bad as that for a surface mount plastic device. > Thats true humidity within plastic housing is changed largely by soldering. This gives more stress on the die than on hermetically devices. But why are the instrument manufacturers using sockets for the LM399 (HP34401A) or reference boards for the LTZ1000 for their pre selected parts? >> >> So 100ppm/15 years outside of "lab conditions" (23 deg , constant >> humidity) is something that I would not guarantee without >> re-calibration. > > I had a feeling that would be the answer - though surely humidity > shouldn't be a factor as these are hermetic parts. The questions > remains though, what level might you specify - if you were forced to > come up with a number (ok a guess!) - for non-selected, non-pre-aged > parts after 15years continuous operation without re-calibration? > Obviously this is given the context of the presumably limited numbers > of samples you've tested and I guess you wouldn't have bothered to > further test early rejects. Sorry this may change from lot to lot. From other (non heated) references I have very different results. Especially from devices bought before and after ROHS. And partly even better graded devices (LT1027BCN) behave worse than standard grade devices (LT1027CCN) in my temperature range. With humidity and LM399: this should be no issue for the metal can package of the LM399. Although I have one LM399 (#1) which I desoldered from a board of unknown age that has a correlation to humidity (see attachment). Ageing is also in the range of 12-15 ppm/year average with seasonal modulation which follows humidity. But since this behavior is not typical for my other LM399 references I would justify this sample as "defect". Humidity: even hermetically parts can suffer from humidity. The epoxy board swells if exposed to humidity. This creates stress to the leads of the package and then to the die. I measure around 0.5ppm/% for plastic parts. In a early publication LT specced around 12ppm change for a 30% rH change for the LT1236AILS8 hermetically package. This spec is now removed from newer data sheets. You will only find a hint to avoid board stress. >> Although typical drift of pre-aged + selected references will be in >> the 1-2ppm/year range if properly treated. > > What would you classify as pre-aged? Do they need to be powered up or > can they be maintained at a suitable temperature? How many rejects > would you expect to get to get one that achieves 1-2ppm? > > Is it known if the major instrument manufacturers preselect and > burn-in LM399s themselves for their middle-range instruments? I'm > pretty sure the top end kit will be all use carefully tested and > selected parts, but what about a 34401A for example? The basic > accuracy spec for that is 20ppm for 90 days, 35ppm for 12 months so > even a 20ppm guaranteed part wouldn't be good enough, especially > allowing margin for drift in other components. I guess I just answered > my own question! > From Keithley cal lab you can see that they adjust the instrument during calibration if they fall outside the 70% window. So I guess that after 1 year they think that ageing of the components (including the reference) is below 30% of the 35 ppm spec per year. The part numbers of the LM399 are not the original ones but a own manufacturer specific. So they do a pre-ageing. Of course powered up. (perhaps intermittend like on resistors?) > I just came across another part which looks very interesting given its > low cost - the automotive qualified REF5050-Q1. Although its only > spec'd as 3ppm/C typical, 8ppm/C max, that's using the box method over > -40 to +125C. The 'typical' chart however, figure 4, page 5 shows the > gradients to be very flat between 25 and 50. Its typical of course, so > real parts may be very different aka Vishay foil resistors. The 0 to > 85C histogram, fig 1 on page 5, do show the majority of parts being in > the range .75ppm/C to 1.75ppm/C which is pretty good, and with luck, > in the 25 to 50C range may well be much better so a crude heating > arrangement may be worthwhile (made easier by the 5050's temperature > output!) > > I can't reconcile fig 4 with the histograms though; from the chart I > reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40 to > 125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show modal > values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong or are > these specs inconsistent? > > Even more surprising is the headline feature on page 1: > > "EXCELLENT LONG-TERM STABILITY: > – 5 ppm/1000 hr (typ) after 1000 hours" > > Unfortunately that seems to be an error as the 'typical' spec on page > 4 is: > 90ppm (0-1000 hours) > 10ppm (1000 to 2000 hours). > > The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96 parts > show the worst case being +25ppm, with the bulk ending approx between > 0 and 15ppm. I wonder if they carry on improving after 2kHrs? > > That's definitely not the SQRT(1kHr) characteristic and is very > different from the standard REF5050 which quotes 100ppm (1st > 100hours), 50ppm (1000 to 2000 hours). > > If you are in a position to pre-age them for 1000 hours that 10ppm > spec is almost as good as the LM399 and best of all, TI quote a price > of $1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from Linear. > One off prices are rather more at $4.15 from Digikey (part no > REF5050AQDRQ1) but again is still a lot cheaper than an LM399 at > $9.92. At $1.60 and .8mA supply current, using 4, 8 or even dozens is > a realistic proposition to exploit statistical improvements and noise > reduction. > > Noise is a bit high at 15uVpp. They're also trimmable. Shame there > isn't an hermetic part though. > > Anybody tried these or spotted the gotchas? Alternatively has anyone > here evaluated the hermetic LTC6655 for long term drift? > > Tony H > Do you really consider unheated references for a 100 ppm lifetime spec? And surface mount SO-8 devices which have a factor 3 worse data on hysteresis (and probably humidity) than the DIP-8 package? I do not know where they get their "typical" T.C. curves with flat area near 25 degrees. Usually they are only measuring 3 points of the references. (-40, 25, +125 deg C or whatever the temperature range of the device is). From theory any bandgap device (REF5050, LTC6655) has to have worse ageing than a buried zener (LM399, AD586). But also the manufacturer specific packaging has its influence. The specs in the datasheets get better the later the datasheets are created for several reasons: The instruments nowadays are with less tolerances. The silicon for the dies is more pure than the elder qualified devices. And of course then marketing demands that every generation has to be better than the previous one. But this has nothing to do with physics and the actual behaviour of the devices. I have my first sample of LTC6655AILS8 measured for T.C. this week. But for my needs it has a too large hysteresis. With best regards Andreas
T
Tony
Fri, Sep 12, 2014 4:46 PM

On 12/09/2014 06:16, Andreas Jahn wrote:

Am 12.09.2014 um 01:06 schrieb Tony Holt:

All ageing specs are "typical" if you want to have "guaranteed"
values you will have to measure it over a reasonable time. (I
recommend min 6 months).
Every treatment (soldering, mechanical/temperature shock) of a
reference may create a new ageing cycle with different slope.

True. I guess that the new ageing cycle from soldering in an LM399 is
not going to be as bad as that for a surface mount plastic device.

Thats true humidity within plastic housing is changed largely by
soldering. This gives more stress on the die than on hermetically
devices.
But why are the instrument manufacturers using sockets for the LM399
(HP34401A) or reference boards for the LTZ1000 for their pre selected
parts?

Well its a lot cheaper to burn-in a device or small reference board than
the whole instrument PCB which is why I expect to do the same. The LM399
on the Keithley 2015 isn't socketed though.

So 100ppm/15 years outside of "lab conditions" (23 deg , constant
humidity) is something that I would not guarantee without
re-calibration.

I had a feeling that would be the answer - though surely humidity
shouldn't be a factor as these are hermetic parts. The questions
remains though, what level might you specify - if you were forced to
come up with a number (ok a guess!) - for non-selected, non-pre-aged
parts after 15years continuous operation without re-calibration?
Obviously this is given the context of the presumably limited numbers
of samples you've tested and I guess you wouldn't have bothered to
further test early rejects.

Sorry this may change from lot to lot. From other (non heated)
references I have very different results. Especially from devices
bought before and after ROHS. And partly even better graded devices
(LT1027BCN) behave worse than standard grade devices (LT1027CCN) in my
temperature range.
With humidity and LM399: this should be no issue for the metal can
package of the LM399. Although I have one LM399 (#1) which I
desoldered from a board of unknown age that has a correlation to
humidity (see attachment). Ageing is also in the range of 12-15
ppm/year average with seasonal modulation which follows humidity. But
since this behavior is not typical for my other LM399 references I
would justify this sample as "defect".

Fair enough. I guess only the reference manufacturers are going to have
any worthwhile long term drift data on statistically large numbers of
devices - apart from customers who aren't going to share the data.

Humidity: even hermetically parts can suffer from humidity. The epoxy
board swells if exposed to humidity. This creates stress to the leads
of the package and then to the die. I measure around 0.5ppm/% for
plastic parts. In a early publication LT specced around 12ppm change
for a 30% rH change for the LT1236AILS8 hermetically package. This
spec is now removed from newer data sheets. You will only find a hint
to avoid board stress.

Yikes! 12ppm for only 30% rh change in a hermetic package?

Although typical drift of pre-aged + selected references will be in
the 1-2ppm/year range if properly treated.

What would you classify as pre-aged? Do they need to be powered up or
can they be maintained at a suitable temperature? How many rejects
would you expect to get to get one that achieves 1-2ppm?

Is it known if the major instrument manufacturers preselect and
burn-in LM399s themselves for their middle-range instruments? I'm
pretty sure the top end kit will be all use carefully tested and
selected parts, but what about a 34401A for example? The basic
accuracy spec for that is 20ppm for 90 days, 35ppm for 12 months so
even a 20ppm guaranteed part wouldn't be good enough, especially
allowing margin for drift in other components. I guess I just
answered my own question!

From Keithley cal lab you can see that they adjust the instrument
during calibration if they fall outside the 70% window. So I guess
that after 1 year they think that ageing of the components (including
the reference) is below 30% of the 35 ppm spec per year. The part
numbers of the LM399 are not the original ones but a own manufacturer
specific.
So they do a pre-ageing. Of course powered up. (perhaps intermittend
like on resistors?)

Interesting. And pre-selected no doubt. I wonder though if heat soaking
at say 40C is almost, if not quite identical to powered up pre-ageing?
If soldered to a board then its a bit different, as in normal operation
the device is going to be warmer than the PCB. For the lower power
band-gap type SMT references though I doubt that having it powered up is
going to make any difference (assuming there aren't any hot components
nearby stressing the PCB). I guess there's plenty of research out there
on this subject.

I just came across another part which looks very interesting given
its low cost - the automotive qualified REF5050-Q1. Although its only
spec'd as 3ppm/C typical, 8ppm/C max, that's using the box method
over -40 to +125C. The 'typical' chart however, figure 4, page 5
shows the gradients to be very flat between 25 and 50. Its typical of
course, so real parts may be very different aka Vishay foil
resistors. The 0 to 85C histogram, fig 1 on page 5, do show the
majority of parts being in the range .75ppm/C to 1.75ppm/C which is
pretty good, and with luck, in the 25 to 50C range may well be much
better so a crude heating arrangement may be worthwhile (made easier
by the 5050's temperature output!)

I can't reconcile fig 4 with the histograms though; from the chart I
reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40 to
125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show
modal values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong or
are these specs inconsistent?

Even more surprising is the headline feature on page 1:

"EXCELLENT LONG-TERM STABILITY:
– 5 ppm/1000 hr (typ) after 1000 hours"

Unfortunately that seems to be an error as the 'typical' spec on page
4 is:
90ppm (0-1000 hours)
10ppm (1000 to 2000 hours).

The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96
parts show the worst case being +25ppm, with the bulk ending approx
between 0 and 15ppm. I wonder if they carry on improving after 2kHrs?

That's definitely not the SQRT(1kHr) characteristic and is very
different from the standard REF5050 which quotes 100ppm (1st
100hours), 50ppm (1000 to 2000 hours).

If you are in a position to pre-age them for 1000 hours that 10ppm
spec is almost as good as the LM399 and best of all, TI quote a price
of $1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from Linear.
One off prices are rather more at $4.15 from Digikey (part no
REF5050AQDRQ1) but again is still a lot cheaper than an LM399 at
$9.92. At $1.60 and .8mA supply current, using 4, 8 or even dozens is
a realistic proposition to exploit statistical improvements and noise
reduction.

Noise is a bit high at 15uVpp. They're also trimmable. Shame there
isn't an hermetic part though.

Anybody tried these or spotted the gotchas? Alternatively has anyone
here evaluated the hermetic LTC6655 for long term drift?

Tony H

Do you really consider unheated references for a 100 ppm lifetime spec?

I'm not confident that 100ppm is realistic for < $10 after your
responses, though I had thought that the LM399 might just achieve it. I
didn't say I would use it unheated - if the drift over 0-40C is too
great, despite selection, I could locate a heater resistor (or two) next
to the reference. The reference has a temperature output and I have an
ADC input (and enough power, hopefully not more than a few hundred
millwatts - not done any experiments though) available so it should be
fairly trivial to keep it within say +/-5C or better. Selecting parts
better than 3ppm/C gives an error budget of ±15ppm or less. The real
killer is long term drift and this part is the first I've seen spec'd at
<= 10ppm (albeit 2nd 1000 hours ageing) apart from expensive metal can
parts.

Also at $1.6 I could afford to do some pre-aging and selection which I
couldn't with those $5+ devices including the LM399 and hermetic LTC6655.

And surface mount SO-8 devices which have a factor 3 worse data on
hysteresis (and probably humidity) than the DIP-8 package?

Hysteresis is spec'd at 10ppm first cycle, 5ppm 2nd cycle, so not a big
problem. They are typical figures of course so actual budget will need
to be somewhat bigger. Since I have a microcontroller which could be
measuring it's temperature continuously, there is the possibility to
correct some of the hysteresis and even the temperature drift if it can
be characterised well enough.

Humidity sensitivity is almost never published, so this is something to
think about. Do you have any suggestions as to how much to budget for an
SO-8 part?

Did you have a particular reference in mind in a DIP-8 package?

I do not know where they get their "typical" T.C. curves with flat
area near 25 degrees. Usually they are only measuring 3 points of the
references. (-40, 25, +125 deg C or whatever the temperature range of
the device is).

They may well only measure 3 points of each device or sample batches for
QC in production, but the device will almost certainly have been very
thoroughly characterised when it was developed, including that
double-humped temperature/voltage change curve in the datasheet. How
well the bulk production parts still match that requires experimentation
but I think its reasonable to believe they will exhibit the same curve
shape.

From theory any bandgap device (REF5050, LTC6655) has to have worse
ageing than a buried zener (LM399, AD586). But also the manufacturer
specific packaging has its influence. The specs in the datasheets get
better the later the datasheets are created for several reasons: The
instruments nowadays are with less tolerances. The silicon for the
dies is more pure than the elder qualified devices. And of course then
marketing demands that every generation has to be better than the
previous one. But this has nothing to do with physics and the actual
behaviour of the devices.

I've seen my fair share of 'creative' or just plain sloppy datasheets,
but I'm not sure its fair to say that datasheets are sailing ever closer
to the wind. I have seen a number of posts alleging that some recent
op-amp datasheets from TI are very poor and a sad reflection of those
from the Nat-Semi era.

It does seem that there hasn't been much progress in precision analogue
components that push the boundaries, or even negative progress as good
components are obsoleted - at least from the mass market manufacturers.
I find it hard to believe that there's still nothing to touch the
LTZ1000 after all these decades - or at least a part that approaches its
performance but is cheaper to buy and use. No doubt due to more and more
being done in the digital domain.

I have my first sample of LTC6655AILS8 measured for T.C. this week.
But for my needs it has a too large hysteresis.

With best regards

Andreas

Thanks,
Tony H

On 12/09/2014 06:16, Andreas Jahn wrote: > > Am 12.09.2014 um 01:06 schrieb Tony Holt: >> >>> All ageing specs are "typical" if you want to have "guaranteed" >>> values you will have to measure it over a reasonable time. (I >>> recommend min 6 months). >>> Every treatment (soldering, mechanical/temperature shock) of a >>> reference may create a new ageing cycle with different slope. >> >> True. I guess that the new ageing cycle from soldering in an LM399 is >> not going to be as bad as that for a surface mount plastic device. >> > Thats true humidity within plastic housing is changed largely by > soldering. This gives more stress on the die than on hermetically > devices. > But why are the instrument manufacturers using sockets for the LM399 > (HP34401A) or reference boards for the LTZ1000 for their pre selected > parts? Well its a lot cheaper to burn-in a device or small reference board than the whole instrument PCB which is why I expect to do the same. The LM399 on the Keithley 2015 isn't socketed though. >>> >>> So 100ppm/15 years outside of "lab conditions" (23 deg , constant >>> humidity) is something that I would not guarantee without >>> re-calibration. >> >> I had a feeling that would be the answer - though surely humidity >> shouldn't be a factor as these are hermetic parts. The questions >> remains though, what level might you specify - if you were forced to >> come up with a number (ok a guess!) - for non-selected, non-pre-aged >> parts after 15years continuous operation without re-calibration? >> Obviously this is given the context of the presumably limited numbers >> of samples you've tested and I guess you wouldn't have bothered to >> further test early rejects. > Sorry this may change from lot to lot. From other (non heated) > references I have very different results. Especially from devices > bought before and after ROHS. And partly even better graded devices > (LT1027BCN) behave worse than standard grade devices (LT1027CCN) in my > temperature range. > With humidity and LM399: this should be no issue for the metal can > package of the LM399. Although I have one LM399 (#1) which I > desoldered from a board of unknown age that has a correlation to > humidity (see attachment). Ageing is also in the range of 12-15 > ppm/year average with seasonal modulation which follows humidity. But > since this behavior is not typical for my other LM399 references I > would justify this sample as "defect". > Fair enough. I guess only the reference manufacturers are going to have any worthwhile long term drift data on statistically large numbers of devices - apart from customers who aren't going to share the data. > Humidity: even hermetically parts can suffer from humidity. The epoxy > board swells if exposed to humidity. This creates stress to the leads > of the package and then to the die. I measure around 0.5ppm/% for > plastic parts. In a early publication LT specced around 12ppm change > for a 30% rH change for the LT1236AILS8 hermetically package. This > spec is now removed from newer data sheets. You will only find a hint > to avoid board stress. > Yikes! 12ppm for only 30% rh change in a hermetic package? >>> Although typical drift of pre-aged + selected references will be in >>> the 1-2ppm/year range if properly treated. >> >> What would you classify as pre-aged? Do they need to be powered up or >> can they be maintained at a suitable temperature? How many rejects >> would you expect to get to get one that achieves 1-2ppm? >> >> Is it known if the major instrument manufacturers preselect and >> burn-in LM399s themselves for their middle-range instruments? I'm >> pretty sure the top end kit will be all use carefully tested and >> selected parts, but what about a 34401A for example? The basic >> accuracy spec for that is 20ppm for 90 days, 35ppm for 12 months so >> even a 20ppm guaranteed part wouldn't be good enough, especially >> allowing margin for drift in other components. I guess I just >> answered my own question! >> > From Keithley cal lab you can see that they adjust the instrument > during calibration if they fall outside the 70% window. So I guess > that after 1 year they think that ageing of the components (including > the reference) is below 30% of the 35 ppm spec per year. The part > numbers of the LM399 are not the original ones but a own manufacturer > specific. > So they do a pre-ageing. Of course powered up. (perhaps intermittend > like on resistors?) Interesting. And pre-selected no doubt. I wonder though if heat soaking at say 40C is almost, if not quite identical to powered up pre-ageing? If soldered to a board then its a bit different, as in normal operation the device is going to be warmer than the PCB. For the lower power band-gap type SMT references though I doubt that having it powered up is going to make any difference (assuming there aren't any hot components nearby stressing the PCB). I guess there's plenty of research out there on this subject. > >> I just came across another part which looks very interesting given >> its low cost - the automotive qualified REF5050-Q1. Although its only >> spec'd as 3ppm/C typical, 8ppm/C max, that's using the box method >> over -40 to +125C. The 'typical' chart however, figure 4, page 5 >> shows the gradients to be very flat between 25 and 50. Its typical of >> course, so real parts may be very different aka Vishay foil >> resistors. The 0 to 85C histogram, fig 1 on page 5, do show the >> majority of parts being in the range .75ppm/C to 1.75ppm/C which is >> pretty good, and with luck, in the 25 to 50C range may well be much >> better so a crude heating arrangement may be worthwhile (made easier >> by the 5050's temperature output!) >> >> I can't reconcile fig 4 with the histograms though; from the chart I >> reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40 to >> 125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show >> modal values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong or >> are these specs inconsistent? >> >> Even more surprising is the headline feature on page 1: >> >> "EXCELLENT LONG-TERM STABILITY: >> – 5 ppm/1000 hr (typ) after 1000 hours" >> >> Unfortunately that seems to be an error as the 'typical' spec on page >> 4 is: >> 90ppm (0-1000 hours) >> 10ppm (1000 to 2000 hours). >> >> The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96 >> parts show the worst case being +25ppm, with the bulk ending approx >> between 0 and 15ppm. I wonder if they carry on improving after 2kHrs? >> >> That's definitely not the SQRT(1kHr) characteristic and is very >> different from the standard REF5050 which quotes 100ppm (1st >> 100hours), 50ppm (1000 to 2000 hours). >> >> If you are in a position to pre-age them for 1000 hours that 10ppm >> spec is almost as good as the LM399 and best of all, TI quote a price >> of $1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from Linear. >> One off prices are rather more at $4.15 from Digikey (part no >> REF5050AQDRQ1) but again is still a lot cheaper than an LM399 at >> $9.92. At $1.60 and .8mA supply current, using 4, 8 or even dozens is >> a realistic proposition to exploit statistical improvements and noise >> reduction. >> >> Noise is a bit high at 15uVpp. They're also trimmable. Shame there >> isn't an hermetic part though. >> >> Anybody tried these or spotted the gotchas? Alternatively has anyone >> here evaluated the hermetic LTC6655 for long term drift? >> >> Tony H >> > Do you really consider unheated references for a 100 ppm lifetime spec? I'm not confident that 100ppm is realistic for < $10 after your responses, though I had thought that the LM399 might just achieve it. I didn't say I would use it unheated - if the drift over 0-40C is too great, despite selection, I could locate a heater resistor (or two) next to the reference. The reference has a temperature output and I have an ADC input (and enough power, hopefully not more than a few hundred millwatts - not done any experiments though) available so it should be fairly trivial to keep it within say +/-5C or better. Selecting parts better than 3ppm/C gives an error budget of ±15ppm or less. The real killer is long term drift and this part is the first I've seen spec'd at <= 10ppm (albeit 2nd 1000 hours ageing) apart from expensive metal can parts. Also at $1.6 I could afford to do some pre-aging and selection which I couldn't with those $5+ devices including the LM399 and hermetic LTC6655. > And surface mount SO-8 devices which have a factor 3 worse data on > hysteresis (and probably humidity) than the DIP-8 package? Hysteresis is spec'd at 10ppm first cycle, 5ppm 2nd cycle, so not a big problem. They are typical figures of course so actual budget will need to be somewhat bigger. Since I have a microcontroller which could be measuring it's temperature continuously, there is the possibility to correct some of the hysteresis and even the temperature drift if it can be characterised well enough. Humidity sensitivity is almost never published, so this is something to think about. Do you have any suggestions as to how much to budget for an SO-8 part? Did you have a particular reference in mind in a DIP-8 package? > I do not know where they get their "typical" T.C. curves with flat > area near 25 degrees. Usually they are only measuring 3 points of the > references. (-40, 25, +125 deg C or whatever the temperature range of > the device is). They may well only measure 3 points of each device or sample batches for QC in production, but the device will almost certainly have been very thoroughly characterised when it was developed, including that double-humped temperature/voltage change curve in the datasheet. How well the bulk production parts still match that requires experimentation but I think its reasonable to believe they will exhibit the same curve shape. > From theory any bandgap device (REF5050, LTC6655) has to have worse > ageing than a buried zener (LM399, AD586). But also the manufacturer > specific packaging has its influence. The specs in the datasheets get > better the later the datasheets are created for several reasons: The > instruments nowadays are with less tolerances. The silicon for the > dies is more pure than the elder qualified devices. And of course then > marketing demands that every generation has to be better than the > previous one. But this has nothing to do with physics and the actual > behaviour of the devices. I've seen my fair share of 'creative' or just plain sloppy datasheets, but I'm not sure its fair to say that datasheets are sailing ever closer to the wind. I have seen a number of posts alleging that some recent op-amp datasheets from TI are very poor and a sad reflection of those from the Nat-Semi era. It does seem that there hasn't been much progress in precision analogue components that push the boundaries, or even negative progress as good components are obsoleted - at least from the mass market manufacturers. I find it hard to believe that there's still nothing to touch the LTZ1000 after all these decades - or at least a part that approaches its performance but is cheaper to buy and use. No doubt due to more and more being done in the digital domain. > I have my first sample of LTC6655AILS8 measured for T.C. this week. > But for my needs it has a too large hysteresis. > > With best regards > > Andreas > Thanks, Tony H
AJ
Andreas Jahn
Sun, Sep 14, 2014 5:29 PM

By the way:

By accident I have found some ageing drift data of REF5050 on Malone's
homepage:
http://www.voltagestandard.com/Tech_Data.html
For me it looks around 120 ppm drift within the first year.
Unfortunately there is no humidity data with the plot.

With best regards

Andreas

Am 12.09.2014 um 18:46 schrieb Tony:

On 12/09/2014 06:16, Andreas Jahn wrote:

Am 12.09.2014 um 01:06 schrieb Tony Holt:

All ageing specs are "typical" if you want to have "guaranteed"
values you will have to measure it over a reasonable time. (I
recommend min 6 months).
Every treatment (soldering, mechanical/temperature shock) of a
reference may create a new ageing cycle with different slope.

True. I guess that the new ageing cycle from soldering in an LM399
is not going to be as bad as that for a surface mount plastic device.

Thats true humidity within plastic housing is changed largely by
soldering. This gives more stress on the die than on hermetically
devices.
But why are the instrument manufacturers using sockets for the LM399
(HP34401A) or reference boards for the LTZ1000 for their pre selected
parts?

Well its a lot cheaper to burn-in a device or small reference board
than the whole instrument PCB which is why I expect to do the same.
The LM399 on the Keithley 2015 isn't socketed though.

So 100ppm/15 years outside of "lab conditions" (23 deg , constant
humidity) is something that I would not guarantee without
re-calibration.

I had a feeling that would be the answer - though surely humidity
shouldn't be a factor as these are hermetic parts. The questions
remains though, what level might you specify - if you were forced to
come up with a number (ok a guess!) - for non-selected, non-pre-aged
parts after 15years continuous operation without re-calibration?
Obviously this is given the context of the presumably limited
numbers of samples you've tested and I guess you wouldn't have
bothered to further test early rejects.

Sorry this may change from lot to lot. From other (non heated)
references I have very different results. Especially from devices
bought before and after ROHS. And partly even better graded devices
(LT1027BCN) behave worse than standard grade devices (LT1027CCN) in
my temperature range.
With humidity and LM399: this should be no issue for the metal can
package of the LM399. Although I have one LM399 (#1) which I
desoldered from a board of unknown age that has a correlation to
humidity (see attachment). Ageing is also in the range of 12-15
ppm/year average with seasonal modulation which follows humidity. But
since this behavior is not typical for my other LM399 references I
would justify this sample as "defect".

Fair enough. I guess only the reference manufacturers are going to
have any worthwhile long term drift data on statistically large
numbers of devices - apart from customers who aren't going to share
the data.

Humidity: even hermetically parts can suffer from humidity. The epoxy
board swells if exposed to humidity. This creates stress to the leads
of the package and then to the die. I measure around 0.5ppm/% for
plastic parts. In a early publication LT specced around 12ppm change
for a 30% rH change for the LT1236AILS8 hermetically package. This
spec is now removed from newer data sheets. You will only find a hint
to avoid board stress.

Yikes! 12ppm for only 30% rh change in a hermetic package?

Although typical drift of pre-aged + selected references will be in
the 1-2ppm/year range if properly treated.

What would you classify as pre-aged? Do they need to be powered up
or can they be maintained at a suitable temperature? How many
rejects would you expect to get to get one that achieves 1-2ppm?

Is it known if the major instrument manufacturers preselect and
burn-in LM399s themselves for their middle-range instruments? I'm
pretty sure the top end kit will be all use carefully tested and
selected parts, but what about a 34401A for example? The basic
accuracy spec for that is 20ppm for 90 days, 35ppm for 12 months so
even a 20ppm guaranteed part wouldn't be good enough, especially
allowing margin for drift in other components. I guess I just
answered my own question!

From Keithley cal lab you can see that they adjust the instrument
during calibration if they fall outside the 70% window. So I guess
that after 1 year they think that ageing of the components (including
the reference) is below 30% of the 35 ppm spec per year. The part
numbers of the LM399 are not the original ones but a own manufacturer
specific.
So they do a pre-ageing. Of course powered up. (perhaps intermittend
like on resistors?)

Interesting. And pre-selected no doubt. I wonder though if heat
soaking at say 40C is almost, if not quite identical to powered up
pre-ageing? If soldered to a board then its a bit different, as in
normal operation the device is going to be warmer than the PCB. For
the lower power band-gap type SMT references though I doubt that
having it powered up is going to make any difference (assuming there
aren't any hot components nearby stressing the PCB). I guess there's
plenty of research out there on this subject.

I just came across another part which looks very interesting given
its low cost - the automotive qualified REF5050-Q1. Although its
only spec'd as 3ppm/C typical, 8ppm/C max, that's using the box
method over -40 to +125C. The 'typical' chart however, figure 4,
page 5 shows the gradients to be very flat between 25 and 50. Its
typical of course, so real parts may be very different aka Vishay
foil resistors. The 0 to 85C histogram, fig 1 on page 5, do show the
majority of parts being in the range .75ppm/C to 1.75ppm/C which is
pretty good, and with luck, in the 25 to 50C range may well be much
better so a crude heating arrangement may be worthwhile (made easier
by the 5050's temperature output!)

I can't reconcile fig 4 with the histograms though; from the chart I
reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40
to 125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show
modal values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong
or are these specs inconsistent?

Even more surprising is the headline feature on page 1:

"EXCELLENT LONG-TERM STABILITY:
– 5 ppm/1000 hr (typ) after 1000 hours"

Unfortunately that seems to be an error as the 'typical' spec on
page 4 is:
90ppm (0-1000 hours)
10ppm (1000 to 2000 hours).

The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96
parts show the worst case being +25ppm, with the bulk ending approx
between 0 and 15ppm. I wonder if they carry on improving after 2kHrs?

That's definitely not the SQRT(1kHr) characteristic and is very
different from the standard REF5050 which quotes 100ppm (1st
100hours), 50ppm (1000 to 2000 hours).

If you are in a position to pre-age them for 1000 hours that 10ppm
spec is almost as good as the LM399 and best of all, TI quote a
price of $1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from
Linear. One off prices are rather more at $4.15 from Digikey (part
no REF5050AQDRQ1) but again is still a lot cheaper than an LM399 at
$9.92. At $1.60 and .8mA supply current, using 4, 8 or even dozens
is a realistic proposition to exploit statistical improvements and
noise reduction.

Noise is a bit high at 15uVpp. They're also trimmable. Shame there
isn't an hermetic part though.

Anybody tried these or spotted the gotchas? Alternatively has anyone
here evaluated the hermetic LTC6655 for long term drift?

Tony H

Do you really consider unheated references for a 100 ppm lifetime spec?

I'm not confident that 100ppm is realistic for < $10 after your
responses, though I had thought that the LM399 might just achieve it.
I didn't say I would use it unheated - if the drift over 0-40C is too
great, despite selection, I could locate a heater resistor (or two)
next to the reference. The reference has a temperature output and I
have an ADC input (and enough power, hopefully not more than a few
hundred millwatts - not done any experiments though) available so it
should be fairly trivial to keep it within say +/-5C or better.
Selecting parts better than 3ppm/C gives an error budget of ±15ppm or
less. The real killer is long term drift and this part is the first
I've seen spec'd at <= 10ppm (albeit 2nd 1000 hours ageing) apart from
expensive metal can parts.

Also at $1.6 I could afford to do some pre-aging and selection which I
couldn't with those $5+ devices including the LM399 and hermetic LTC6655.

And surface mount SO-8 devices which have a factor 3 worse data on
hysteresis (and probably humidity) than the DIP-8 package?

Hysteresis is spec'd at 10ppm first cycle, 5ppm 2nd cycle, so not a
big problem. They are typical figures of course so actual budget will
need to be somewhat bigger. Since I have a microcontroller which could
be measuring it's temperature continuously, there is the possibility
to correct some of the hysteresis and even the temperature drift if it
can be characterised well enough.

Humidity sensitivity is almost never published, so this is something
to think about. Do you have any suggestions as to how much to budget
for an SO-8 part?

Did you have a particular reference in mind in a DIP-8 package?

I do not know where they get their "typical" T.C. curves with flat
area near 25 degrees. Usually they are only measuring 3 points of the
references. (-40, 25, +125 deg C or whatever the temperature range of
the device is).

They may well only measure 3 points of each device or sample batches
for QC in production, but the device will almost certainly have been
very thoroughly characterised when it was developed, including that
double-humped temperature/voltage change curve in the datasheet. How
well the bulk production parts still match that requires
experimentation but I think its reasonable to believe they will
exhibit the same curve shape.

From theory any bandgap device (REF5050, LTC6655) has to have worse
ageing than a buried zener (LM399, AD586). But also the manufacturer
specific packaging has its influence. The specs in the datasheets get
better the later the datasheets are created for several reasons: The
instruments nowadays are with less tolerances. The silicon for the
dies is more pure than the elder qualified devices. And of course
then marketing demands that every generation has to be better than
the previous one. But this has nothing to do with physics and the
actual behaviour of the devices.

I've seen my fair share of 'creative' or just plain sloppy datasheets,
but I'm not sure its fair to say that datasheets are sailing ever
closer to the wind. I have seen a number of posts alleging that some
recent op-amp datasheets from TI are very poor and a sad reflection of
those from the Nat-Semi era.

It does seem that there hasn't been much progress in precision
analogue components that push the boundaries, or even negative
progress as good components are obsoleted - at least from the mass
market manufacturers. I find it hard to believe that there's still
nothing to touch the LTZ1000 after all these decades - or at least a
part that approaches its performance but is cheaper to buy and use. No
doubt due to more and more being done in the digital domain.

I have my first sample of LTC6655AILS8 measured for T.C. this week.
But for my needs it has a too large hysteresis.

With best regards

Andreas

Thanks,
Tony H


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

By the way: By accident I have found some ageing drift data of REF5050 on Malone's homepage: http://www.voltagestandard.com/Tech_Data.html For me it looks around 120 ppm drift within the first year. Unfortunately there is no humidity data with the plot. With best regards Andreas Am 12.09.2014 um 18:46 schrieb Tony: > On 12/09/2014 06:16, Andreas Jahn wrote: >> >> Am 12.09.2014 um 01:06 schrieb Tony Holt: >>> >>>> All ageing specs are "typical" if you want to have "guaranteed" >>>> values you will have to measure it over a reasonable time. (I >>>> recommend min 6 months). >>>> Every treatment (soldering, mechanical/temperature shock) of a >>>> reference may create a new ageing cycle with different slope. >>> >>> True. I guess that the new ageing cycle from soldering in an LM399 >>> is not going to be as bad as that for a surface mount plastic device. >>> >> Thats true humidity within plastic housing is changed largely by >> soldering. This gives more stress on the die than on hermetically >> devices. >> But why are the instrument manufacturers using sockets for the LM399 >> (HP34401A) or reference boards for the LTZ1000 for their pre selected >> parts? > > Well its a lot cheaper to burn-in a device or small reference board > than the whole instrument PCB which is why I expect to do the same. > The LM399 on the Keithley 2015 isn't socketed though. > >>>> >>>> So 100ppm/15 years outside of "lab conditions" (23 deg , constant >>>> humidity) is something that I would not guarantee without >>>> re-calibration. >>> >>> I had a feeling that would be the answer - though surely humidity >>> shouldn't be a factor as these are hermetic parts. The questions >>> remains though, what level might you specify - if you were forced to >>> come up with a number (ok a guess!) - for non-selected, non-pre-aged >>> parts after 15years continuous operation without re-calibration? >>> Obviously this is given the context of the presumably limited >>> numbers of samples you've tested and I guess you wouldn't have >>> bothered to further test early rejects. >> Sorry this may change from lot to lot. From other (non heated) >> references I have very different results. Especially from devices >> bought before and after ROHS. And partly even better graded devices >> (LT1027BCN) behave worse than standard grade devices (LT1027CCN) in >> my temperature range. >> With humidity and LM399: this should be no issue for the metal can >> package of the LM399. Although I have one LM399 (#1) which I >> desoldered from a board of unknown age that has a correlation to >> humidity (see attachment). Ageing is also in the range of 12-15 >> ppm/year average with seasonal modulation which follows humidity. But >> since this behavior is not typical for my other LM399 references I >> would justify this sample as "defect". >> > > Fair enough. I guess only the reference manufacturers are going to > have any worthwhile long term drift data on statistically large > numbers of devices - apart from customers who aren't going to share > the data. > >> Humidity: even hermetically parts can suffer from humidity. The epoxy >> board swells if exposed to humidity. This creates stress to the leads >> of the package and then to the die. I measure around 0.5ppm/% for >> plastic parts. In a early publication LT specced around 12ppm change >> for a 30% rH change for the LT1236AILS8 hermetically package. This >> spec is now removed from newer data sheets. You will only find a hint >> to avoid board stress. >> > Yikes! 12ppm for only 30% rh change in a hermetic package? >>>> Although typical drift of pre-aged + selected references will be in >>>> the 1-2ppm/year range if properly treated. >>> >>> What would you classify as pre-aged? Do they need to be powered up >>> or can they be maintained at a suitable temperature? How many >>> rejects would you expect to get to get one that achieves 1-2ppm? >>> >>> Is it known if the major instrument manufacturers preselect and >>> burn-in LM399s themselves for their middle-range instruments? I'm >>> pretty sure the top end kit will be all use carefully tested and >>> selected parts, but what about a 34401A for example? The basic >>> accuracy spec for that is 20ppm for 90 days, 35ppm for 12 months so >>> even a 20ppm guaranteed part wouldn't be good enough, especially >>> allowing margin for drift in other components. I guess I just >>> answered my own question! >>> >> From Keithley cal lab you can see that they adjust the instrument >> during calibration if they fall outside the 70% window. So I guess >> that after 1 year they think that ageing of the components (including >> the reference) is below 30% of the 35 ppm spec per year. The part >> numbers of the LM399 are not the original ones but a own manufacturer >> specific. >> So they do a pre-ageing. Of course powered up. (perhaps intermittend >> like on resistors?) > > Interesting. And pre-selected no doubt. I wonder though if heat > soaking at say 40C is almost, if not quite identical to powered up > pre-ageing? If soldered to a board then its a bit different, as in > normal operation the device is going to be warmer than the PCB. For > the lower power band-gap type SMT references though I doubt that > having it powered up is going to make any difference (assuming there > aren't any hot components nearby stressing the PCB). I guess there's > plenty of research out there on this subject. > >> >>> I just came across another part which looks very interesting given >>> its low cost - the automotive qualified REF5050-Q1. Although its >>> only spec'd as 3ppm/C typical, 8ppm/C max, that's using the box >>> method over -40 to +125C. The 'typical' chart however, figure 4, >>> page 5 shows the gradients to be very flat between 25 and 50. Its >>> typical of course, so real parts may be very different aka Vishay >>> foil resistors. The 0 to 85C histogram, fig 1 on page 5, do show the >>> majority of parts being in the range .75ppm/C to 1.75ppm/C which is >>> pretty good, and with luck, in the 25 to 50C range may well be much >>> better so a crude heating arrangement may be worthwhile (made easier >>> by the 5050's temperature output!) >>> >>> I can't reconcile fig 4 with the histograms though; from the chart I >>> reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40 >>> to 125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show >>> modal values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong >>> or are these specs inconsistent? >>> >>> Even more surprising is the headline feature on page 1: >>> >>> "EXCELLENT LONG-TERM STABILITY: >>> – 5 ppm/1000 hr (typ) after 1000 hours" >>> >>> Unfortunately that seems to be an error as the 'typical' spec on >>> page 4 is: >>> 90ppm (0-1000 hours) >>> 10ppm (1000 to 2000 hours). >>> >>> The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96 >>> parts show the worst case being +25ppm, with the bulk ending approx >>> between 0 and 15ppm. I wonder if they carry on improving after 2kHrs? >>> >>> That's definitely not the SQRT(1kHr) characteristic and is very >>> different from the standard REF5050 which quotes 100ppm (1st >>> 100hours), 50ppm (1000 to 2000 hours). >>> >>> If you are in a position to pre-age them for 1000 hours that 10ppm >>> spec is almost as good as the LM399 and best of all, TI quote a >>> price of $1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from >>> Linear. One off prices are rather more at $4.15 from Digikey (part >>> no REF5050AQDRQ1) but again is still a lot cheaper than an LM399 at >>> $9.92. At $1.60 and .8mA supply current, using 4, 8 or even dozens >>> is a realistic proposition to exploit statistical improvements and >>> noise reduction. >>> >>> Noise is a bit high at 15uVpp. They're also trimmable. Shame there >>> isn't an hermetic part though. >>> >>> Anybody tried these or spotted the gotchas? Alternatively has anyone >>> here evaluated the hermetic LTC6655 for long term drift? >>> >>> Tony H >>> >> Do you really consider unheated references for a 100 ppm lifetime spec? > > I'm not confident that 100ppm is realistic for < $10 after your > responses, though I had thought that the LM399 might just achieve it. > I didn't say I would use it unheated - if the drift over 0-40C is too > great, despite selection, I could locate a heater resistor (or two) > next to the reference. The reference has a temperature output and I > have an ADC input (and enough power, hopefully not more than a few > hundred millwatts - not done any experiments though) available so it > should be fairly trivial to keep it within say +/-5C or better. > Selecting parts better than 3ppm/C gives an error budget of ±15ppm or > less. The real killer is long term drift and this part is the first > I've seen spec'd at <= 10ppm (albeit 2nd 1000 hours ageing) apart from > expensive metal can parts. > > Also at $1.6 I could afford to do some pre-aging and selection which I > couldn't with those $5+ devices including the LM399 and hermetic LTC6655. > >> And surface mount SO-8 devices which have a factor 3 worse data on >> hysteresis (and probably humidity) than the DIP-8 package? > > Hysteresis is spec'd at 10ppm first cycle, 5ppm 2nd cycle, so not a > big problem. They are typical figures of course so actual budget will > need to be somewhat bigger. Since I have a microcontroller which could > be measuring it's temperature continuously, there is the possibility > to correct some of the hysteresis and even the temperature drift if it > can be characterised well enough. > > Humidity sensitivity is almost never published, so this is something > to think about. Do you have any suggestions as to how much to budget > for an SO-8 part? > > Did you have a particular reference in mind in a DIP-8 package? > >> I do not know where they get their "typical" T.C. curves with flat >> area near 25 degrees. Usually they are only measuring 3 points of the >> references. (-40, 25, +125 deg C or whatever the temperature range of >> the device is). > > They may well only measure 3 points of each device or sample batches > for QC in production, but the device will almost certainly have been > very thoroughly characterised when it was developed, including that > double-humped temperature/voltage change curve in the datasheet. How > well the bulk production parts still match that requires > experimentation but I think its reasonable to believe they will > exhibit the same curve shape. > >> From theory any bandgap device (REF5050, LTC6655) has to have worse >> ageing than a buried zener (LM399, AD586). But also the manufacturer >> specific packaging has its influence. The specs in the datasheets get >> better the later the datasheets are created for several reasons: The >> instruments nowadays are with less tolerances. The silicon for the >> dies is more pure than the elder qualified devices. And of course >> then marketing demands that every generation has to be better than >> the previous one. But this has nothing to do with physics and the >> actual behaviour of the devices. > > I've seen my fair share of 'creative' or just plain sloppy datasheets, > but I'm not sure its fair to say that datasheets are sailing ever > closer to the wind. I have seen a number of posts alleging that some > recent op-amp datasheets from TI are very poor and a sad reflection of > those from the Nat-Semi era. > > It does seem that there hasn't been much progress in precision > analogue components that push the boundaries, or even negative > progress as good components are obsoleted - at least from the mass > market manufacturers. I find it hard to believe that there's still > nothing to touch the LTZ1000 after all these decades - or at least a > part that approaches its performance but is cheaper to buy and use. No > doubt due to more and more being done in the digital domain. > >> I have my first sample of LTC6655AILS8 measured for T.C. this week. >> But for my needs it has a too large hysteresis. >> >> With best regards >> >> Andreas >> > Thanks, > Tony H > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
T
Tony
Mon, Sep 15, 2014 4:57 PM

On 14/09/2014 18:29, Andreas Jahn wrote:

By the way:

By accident I have found some ageing drift data of REF5050 on Malone's
homepage:
http://www.voltagestandard.com/Tech_Data.html
For me it looks around 120 ppm drift within the first year.
Unfortunately there is no humidity data with the plot.

With best regards

Andreas

That's a very interesting find Andreas, thanks. After approx 500 hours,
maximum drift is approx 80ppm, but its impossible to deduce how much of
that is due to long term drift, temperature and humidity changes.

The long term drift specification for the REF5050A tested is 125ppm
(0-1000 hours), 45ppm (1000 to 2000 hours). Those for the automotive
qualified REF5050-Q1 part are 100ppm and 10ppm respectively - way better
after the initial 1k hours.

Regards,

Tony H

On 14/09/2014 18:29, Andreas Jahn wrote: > By the way: > > By accident I have found some ageing drift data of REF5050 on Malone's > homepage: > http://www.voltagestandard.com/Tech_Data.html > For me it looks around 120 ppm drift within the first year. > Unfortunately there is no humidity data with the plot. > > With best regards > > Andreas That's a very interesting find Andreas, thanks. After approx 500 hours, maximum drift is approx 80ppm, but its impossible to deduce how much of that is due to long term drift, temperature and humidity changes. The long term drift specification for the REF5050A tested is 125ppm (0-1000 hours), 45ppm (1000 to 2000 hours). Those for the automotive qualified REF5050-Q1 part are 100ppm and 10ppm respectively - way better after the initial 1k hours. Regards, Tony H
MK
M K
Mon, Sep 15, 2014 5:35 PM

On 15/09/2014 17:57, Tony wrote:

On 14/09/2014 18:29, Andreas Jahn wrote:

By the way:

By accident I have found some ageing drift data of REF5050 on
Malone's homepage:
http://www.voltagestandard.com/Tech_Data.html
For me it looks around 120 ppm drift within the first year.
Unfortunately there is no humidity data with the plot.

With best regards

Andreas

That's a very interesting find Andreas, thanks. After approx 500
hours, maximum drift is approx 80ppm, but its impossible to deduce how
much of that is due to long term drift, temperature and humidity changes.

The long term drift specification for the REF5050A tested is 125ppm
(0-1000 hours), 45ppm (1000 to 2000 hours). Those for the automotive
qualified REF5050-Q1 part are 100ppm and 10ppm respectively - way
better after the initial 1k hours.

Regards,

Tony H

One thing that concerns me is that all 5 units seem to drift in sync, so
how much of that is the DVM doing the measuring? or is it just a single
batch with the same date code?

On 15/09/2014 17:57, Tony wrote: > On 14/09/2014 18:29, Andreas Jahn wrote: >> By the way: >> >> By accident I have found some ageing drift data of REF5050 on >> Malone's homepage: >> http://www.voltagestandard.com/Tech_Data.html >> For me it looks around 120 ppm drift within the first year. >> Unfortunately there is no humidity data with the plot. >> >> With best regards >> >> Andreas > That's a very interesting find Andreas, thanks. After approx 500 > hours, maximum drift is approx 80ppm, but its impossible to deduce how > much of that is due to long term drift, temperature and humidity changes. > > The long term drift specification for the REF5050A tested is 125ppm > (0-1000 hours), 45ppm (1000 to 2000 hours). Those for the automotive > qualified REF5050-Q1 part are 100ppm and 10ppm respectively - way > better after the initial 1k hours. > > Regards, > > Tony H One thing that concerns me is that all 5 units seem to drift in sync, so how much of that is the DVM doing the measuring? or is it just a single batch with the same date code?
GA
George Atkinson
Mon, Sep 15, 2014 7:57 PM

Hi,
I'm late to this thread, but have some experience. Several years ago I did
some research on replacing xenon flashlamps in a high speed fluorescence
instrument application. As the white LED candidates had phosphor coatings I
assumed (don't assume - guess ;-) that they would be too slow. It turned
out that they were significantly faster than the xenon lamps, with sharp
turn-off rather than the tail of the flashlamp.
So don't dismiss white LEDs for modulated applications.

Robert G8RPI.

On 11 September 2014 13:14, Chuck Harris cfharris@erols.com wrote:

Hi Dallas,

Let us know if it helps.  Neons are very fast light sources, and
can switch way faster than the CdS photo resistors can respond.

A nice red/orange LED should work well with the original CdS
photo resistors.

-Chuck Harris

Dallas Smith wrote:

Chuck,

After searching for the wave length of NE2 neon bulbs, I found that the
spectrum is
between 600-650nm. So I ordered red-org at 615nm. The white LED's I've
have do have
the characteristics you mentioned, Thank you for all your input, maybe
the results
will show some improvement.

Dallas


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.

Hi, I'm late to this thread, but have some experience. Several years ago I did some research on replacing xenon flashlamps in a high speed fluorescence instrument application. As the white LED candidates had phosphor coatings I assumed (don't assume - guess ;-) that they would be too slow. It turned out that they were significantly faster than the xenon lamps, with sharp turn-off rather than the tail of the flashlamp. So don't dismiss white LEDs for modulated applications. Robert G8RPI. On 11 September 2014 13:14, Chuck Harris <cfharris@erols.com> wrote: > Hi Dallas, > > Let us know if it helps. Neons are very fast light sources, and > can switch way faster than the CdS photo resistors can respond. > > A nice red/orange LED should work well with the original CdS > photo resistors. > > -Chuck Harris > > > Dallas Smith wrote: > >> Chuck, >> >> After searching for the wave length of NE2 neon bulbs, I found that the >> spectrum is >> between 600-650nm. So I ordered red-org at 615nm. The white LED's I've >> have do have >> the characteristics you mentioned, Thank you for all your input, maybe >> the results >> will show some improvement. >> >> Dallas >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/volt-nuts >> and follow the instructions there. >> > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/volt-nuts > and follow the instructions there. >
DS
Dallas Smith
Mon, Sep 15, 2014 8:19 PM

Thank you Robert for you experience,

My white LED's did have a persistence to them, so I'm going to change
them to ORG/RED (615nm). I've temporarily using some low output amber
LED's and determined that the caps I changed needed to be changed back
to original values. I also modified the zero control circuit with a 1.25
volt reference centered at zero with two switching diodes to base the
offset at zero for less drift. I,m considering using the Linear chopper
amp Randy's using. But I'm not sure how to incorporate the gain control
at S101B? The over all feedback should be no problem.

Dallas

On 9/15/2014 3:57 PM, George Atkinson wrote:

Hi,
I'm late to this thread, but have some experience. Several years ago I did
some research on replacing xenon flashlamps in a high speed fluorescence
instrument application. As the white LED candidates had phosphor coatings I
assumed (don't assume - guess ;-) that they would be too slow. It turned
out that they were significantly faster than the xenon lamps, with sharp
turn-off rather than the tail of the flashlamp.
So don't dismiss white LEDs for modulated applications.

Robert G8RPI.

On 11 September 2014 13:14, Chuck Harris cfharris@erols.com wrote:

Hi Dallas,

Let us know if it helps.  Neons are very fast light sources, and
can switch way faster than the CdS photo resistors can respond.

A nice red/orange LED should work well with the original CdS
photo resistors.

-Chuck Harris

Dallas Smith wrote:

Chuck,

After searching for the wave length of NE2 neon bulbs, I found that the
spectrum is
between 600-650nm. So I ordered red-org at 615nm. The white LED's I've
have do have
the characteristics you mentioned, Thank you for all your input, maybe
the results
will show some improvement.

Dallas


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Thank you Robert for you experience, My white LED's did have a persistence to them, so I'm going to change them to ORG/RED (615nm). I've temporarily using some low output amber LED's and determined that the caps I changed needed to be changed back to original values. I also modified the zero control circuit with a 1.25 volt reference centered at zero with two switching diodes to base the offset at zero for less drift. I,m considering using the Linear chopper amp Randy's using. But I'm not sure how to incorporate the gain control at S101B? The over all feedback should be no problem. Dallas On 9/15/2014 3:57 PM, George Atkinson wrote: > Hi, > I'm late to this thread, but have some experience. Several years ago I did > some research on replacing xenon flashlamps in a high speed fluorescence > instrument application. As the white LED candidates had phosphor coatings I > assumed (don't assume - guess ;-) that they would be too slow. It turned > out that they were significantly faster than the xenon lamps, with sharp > turn-off rather than the tail of the flashlamp. > So don't dismiss white LEDs for modulated applications. > > Robert G8RPI. > > On 11 September 2014 13:14, Chuck Harris <cfharris@erols.com> wrote: > >> Hi Dallas, >> >> Let us know if it helps. Neons are very fast light sources, and >> can switch way faster than the CdS photo resistors can respond. >> >> A nice red/orange LED should work well with the original CdS >> photo resistors. >> >> -Chuck Harris >> >> >> Dallas Smith wrote: >> >>> Chuck, >>> >>> After searching for the wave length of NE2 neon bulbs, I found that the >>> spectrum is >>> between 600-650nm. So I ordered red-org at 615nm. The white LED's I've >>> have do have >>> the characteristics you mentioned, Thank you for all your input, maybe >>> the results >>> will show some improvement. >>> >>> Dallas >>> _______________________________________________ >>> volt-nuts mailing list -- volt-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>> mailman/listinfo/volt-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/volt-nuts >> and follow the instructions there. >> > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
AJ
Andreas Jahn
Mon, Sep 15, 2014 8:20 PM

Hello,

the drift in sync is something that I have with humidity changes and
plastic housings.
And they also have temperature changes of +/-4 deg C in the setup.
But usually the temperature gradient is different with different samples
near room temperature.
The influence may also be due to PCB stress.

with best regards

Andreas

Am 15.09.2014 um 19:35 schrieb M K:

On 15/09/2014 17:57, Tony wrote:

On 14/09/2014 18:29, Andreas Jahn wrote:

By the way:

By accident I have found some ageing drift data of REF5050 on
Malone's homepage:
http://www.voltagestandard.com/Tech_Data.html
For me it looks around 120 ppm drift within the first year.
Unfortunately there is no humidity data with the plot.

With best regards

Andreas

That's a very interesting find Andreas, thanks. After approx 500
hours, maximum drift is approx 80ppm, but its impossible to deduce
how much of that is due to long term drift, temperature and humidity
changes.

The long term drift specification for the REF5050A tested is 125ppm
(0-1000 hours), 45ppm (1000 to 2000 hours). Those for the automotive
qualified REF5050-Q1 part are 100ppm and 10ppm respectively - way
better after the initial 1k hours.

Regards,

Tony H

One thing that concerns me is that all 5 units seem to drift in sync,
so how much of that is the DVM doing the measuring? or is it just a
single batch with the same date code?


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hello, the drift in sync is something that I have with humidity changes and plastic housings. And they also have temperature changes of +/-4 deg C in the setup. But usually the temperature gradient is different with different samples near room temperature. The influence may also be due to PCB stress. with best regards Andreas Am 15.09.2014 um 19:35 schrieb M K: > On 15/09/2014 17:57, Tony wrote: >> On 14/09/2014 18:29, Andreas Jahn wrote: >>> By the way: >>> >>> By accident I have found some ageing drift data of REF5050 on >>> Malone's homepage: >>> http://www.voltagestandard.com/Tech_Data.html >>> For me it looks around 120 ppm drift within the first year. >>> Unfortunately there is no humidity data with the plot. >>> >>> With best regards >>> >>> Andreas >> That's a very interesting find Andreas, thanks. After approx 500 >> hours, maximum drift is approx 80ppm, but its impossible to deduce >> how much of that is due to long term drift, temperature and humidity >> changes. >> >> The long term drift specification for the REF5050A tested is 125ppm >> (0-1000 hours), 45ppm (1000 to 2000 hours). Those for the automotive >> qualified REF5050-Q1 part are 100ppm and 10ppm respectively - way >> better after the initial 1k hours. >> >> Regards, >> >> Tony H > One thing that concerns me is that all 5 units seem to drift in sync, > so how much of that is the DVM doing the measuring? or is it just a > single batch with the same date code? > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.