volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

The 3458A A/D

FS
Frank Stellmach
Mon, Jan 20, 2014 11:09 PM

"3. The 3458A still has got the best linear A/D, around 0.02ppm of input."

Is that a typo? Even 0.2 ppm would seem extremely good. I think I saw an
official number of 0.1ppm somewhere. Probably 0.1ppm of range. And how do
they implement that? Manual trimming and compensation?

Hello Jan,

well DNL, INL, sometimes I still struggle with those terms..

Those 0.02ppm were nonlinearity of input, measured with a JJ array.
And they use a special multiple slope scheme, with glitch compensation for the switching FETs.

Just download the April 1989 HP Journal, there's everything explained in detail.

http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf

Frank

>>"3. The 3458A still has got the best linear A/D, around 0.02ppm of input." >Is that a typo? Even 0.2 ppm would seem extremely good. I think I saw an >official number of 0.1ppm somewhere. Probably 0.1ppm of range. And how do >they implement that? Manual trimming and compensation? Hello Jan, well DNL, INL, sometimes I still struggle with those terms.. Those 0.02ppm were nonlinearity of input, measured with a JJ array. And they use a special multiple slope scheme, with glitch compensation for the switching FETs. Just download the April 1989 HP Journal, there's everything explained in detail. http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf Frank
JP
John Phillips
Mon, Jan 20, 2014 11:33 PM

The other reason the price is so high is that the cost of entry for a
competitor to develop and sell a comparable unit for such low sales volume
would also cost that much if not more. The Fluke 8508 does cost more but it
is a 3458A "under the hood".

On Mon, Jan 20, 2014 at 3:09 PM, Frank Stellmach <frank.stellmach@freenet.de

wrote:

"3. The 3458A still has got the best linear A/D, around 0.02ppm of input."

Is that a typo? Even 0.2 ppm would seem extremely good. I think I saw an

official number of 0.1ppm somewhere. Probably 0.1ppm of range. And how do
they implement that? Manual trimming and compensation?

Hello Jan,

well DNL, INL, sometimes I still struggle with those terms..

Those 0.02ppm were nonlinearity of input, measured with a JJ array.
And they use a special multiple slope scheme, with glitch compensation for
the switching FETs.

Just download the April 1989 HP Journal, there's everything explained in
detail.

http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf

Frank


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.

--
John Phillips

The other reason the price is so high is that the cost of entry for a competitor to develop and sell a comparable unit for such low sales volume would also cost that much if not more. The Fluke 8508 does cost more but it is a 3458A "under the hood". On Mon, Jan 20, 2014 at 3:09 PM, Frank Stellmach <frank.stellmach@freenet.de > wrote: > "3. The 3458A still has got the best linear A/D, around 0.02ppm of input." >>> >> > Is that a typo? Even 0.2 ppm would seem extremely good. I think I saw an >> official number of 0.1ppm somewhere. Probably 0.1ppm of range. And how do >> they implement that? Manual trimming and compensation? >> > > Hello Jan, > > well DNL, INL, sometimes I still struggle with those terms.. > > Those 0.02ppm were nonlinearity of input, measured with a JJ array. > And they use a special multiple slope scheme, with glitch compensation for > the switching FETs. > > Just download the April 1989 HP Journal, there's everything explained in > detail. > > http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf > > > Frank > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/volt-nuts > and follow the instructions there. > -- John Phillips
TK
Tom Knox
Tue, Jan 21, 2014 2:04 AM

The Fluke 8508 is actually more of a Wavetek/Datron 1281 under the hood. The real difficulty in introducing a 3458A competitor is that the scientific community has been characterizing the HP meter for three decades. It's quarks and idiosyncrasies are well know such as how it will drift, or react to environmental changes.  Even a superiour meter will take years to be as usable.

Thomas Knox

From: john.phillips0@gmail.com
Date: Mon, 20 Jan 2014 15:33:39 -0800
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] The 3458A A/D

The other reason the price is so high is that the cost of entry for a
competitor to develop and sell a comparable unit for such low sales volume
would also cost that much if not more. The Fluke 8508 does cost more but it
is a 3458A "under the hood".

On Mon, Jan 20, 2014 at 3:09 PM, Frank Stellmach <frank.stellmach@freenet.de

wrote:

"3. The 3458A still has got the best linear A/D, around 0.02ppm of input."

Is that a typo? Even 0.2 ppm would seem extremely good. I think I saw an

official number of 0.1ppm somewhere. Probably 0.1ppm of range. And how do
they implement that? Manual trimming and compensation?

Hello Jan,

well DNL, INL, sometimes I still struggle with those terms..

Those 0.02ppm were nonlinearity of input, measured with a JJ array.
And they use a special multiple slope scheme, with glitch compensation for
the switching FETs.

Just download the April 1989 HP Journal, there's everything explained in
detail.

http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf

Frank


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.

--
John Phillips


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

The Fluke 8508 is actually more of a Wavetek/Datron 1281 under the hood. The real difficulty in introducing a 3458A competitor is that the scientific community has been characterizing the HP meter for three decades. It's quarks and idiosyncrasies are well know such as how it will drift, or react to environmental changes. Even a superiour meter will take years to be as usable. Thomas Knox > From: john.phillips0@gmail.com > Date: Mon, 20 Jan 2014 15:33:39 -0800 > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] The 3458A A/D > > The other reason the price is so high is that the cost of entry for a > competitor to develop and sell a comparable unit for such low sales volume > would also cost that much if not more. The Fluke 8508 does cost more but it > is a 3458A "under the hood". > > > On Mon, Jan 20, 2014 at 3:09 PM, Frank Stellmach <frank.stellmach@freenet.de > > wrote: > > > "3. The 3458A still has got the best linear A/D, around 0.02ppm of input." > >>> > >> > > Is that a typo? Even 0.2 ppm would seem extremely good. I think I saw an > >> official number of 0.1ppm somewhere. Probably 0.1ppm of range. And how do > >> they implement that? Manual trimming and compensation? > >> > > > > Hello Jan, > > > > well DNL, INL, sometimes I still struggle with those terms.. > > > > Those 0.02ppm were nonlinearity of input, measured with a JJ array. > > And they use a special multiple slope scheme, with glitch compensation for > > the switching FETs. > > > > Just download the April 1989 HP Journal, there's everything explained in > > detail. > > > > http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1989-04.pdf > > > > > > Frank > > > > > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > > mailman/listinfo/volt-nuts > > and follow the instructions there. > > > > > > -- > John Phillips > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.