5680A what is a reasonable case temp ?

PL
Pete Lancashire
Tue, Feb 28, 2012 7:24 PM

I'm ready to mount on of my 5680A in a case and looking for suggestions on
what one would consider a reasonable maximum case temp. Not being
a blood sucking cell provider,  I can't replace a dead one every year.

Eventually I see locking the 5680A to GPS if that changes what one would
suggest.

-pete

I'm ready to mount on of my 5680A in a case and looking for suggestions on what one would consider a reasonable maximum case temp. Not being a blood sucking cell provider, I can't replace a dead one every year. Eventually I see locking the 5680A to GPS if that changes what one would suggest. -pete
CA
Chris Albertson
Tue, Feb 28, 2012 7:42 PM

On Tue, Feb 28, 2012 at 11:24 AM, Pete Lancashire
pete@petelancashire.com wrote:

I'm ready to mount on of my 5680A in a case and looking for suggestions on
what one would consider a reasonable maximum case temp. Not being
a blood sucking cell provider,  I can't replace a dead one every year.

I think in practice you are going to need to mount the 5680 to an
aluminum plate at least twice the size of the 5680 and then keep a
moderate airflow over the top and bottom.  Perhaps mount the  plate
using standoffs.  It's clear to me from seeing all those screws that
the 5680 needs to be firmly connected to some kind of heat sink even
it only a large plate.

Eventually I see locking the 5680A to GPS if that changes what one would

Yes you pretty much have to calibrate the 5680.  But what do you mean
"if that changes"?  the 5680 will drift away from any calibration.

Chris Albertson
Redondo Beach, California

On Tue, Feb 28, 2012 at 11:24 AM, Pete Lancashire <pete@petelancashire.com> wrote: > I'm ready to mount on of my 5680A in a case and looking for suggestions on > what one would consider a reasonable maximum case temp. Not being > a blood sucking cell provider,  I can't replace a dead one every year. I think in practice you are going to need to mount the 5680 to an aluminum plate at least twice the size of the 5680 and then keep a moderate airflow over the top and bottom. Perhaps mount the plate using standoffs. It's clear to me from seeing all those screws that the 5680 needs to be firmly connected to some kind of heat sink even it only a large plate. > > Eventually I see locking the 5680A to GPS if that changes what one would Yes you pretty much have to calibrate the 5680. But what do you mean "if that changes"? the 5680 will drift away from any calibration. Chris Albertson Redondo Beach, California
B
beale
Tue, Feb 28, 2012 8:06 PM

From: Chris Albertson albertson.chris@gmail.com
I'm ready to mount on of my 5680A in a case and looking for suggestions on
what one would consider a reasonable maximum case temp.

With the FE-5680A mounted to a 4x6x1" finned aluminum heatsink as shown here:
https://picasaweb.google.com/109928236040342205185/FE5680A#5685636673930493138

the opposite side of the case reaches 55 C in a 20 C ambient, with no fan whatsoever. Running a 40mm 12V fan at 5V aimed at the heatsink drops that to 45 C. At 5V the fan is nearly silent, and the airflow is quite weak.  Despite the internal temperature compensation via DDS offset updates that the unit apparently does, I measured one of my units to have a 7E-12 per degree C temperature sensitivity, larger than I expected. So I do want to use a thermostat on the fan.

The Symmetricom XPRO manual http://www.symmetricom.com/link.cfm?lid=7082  (which I believe is a somewhat similar Rb design) estimates that the lifetime of that unit doubles when baseplate temperature is reduced from 60 C to 40 C (MTBF: 243k hours => 591k hours).  Note that 243k hours is about 28 years.

On the other hand, my assumption is these surplus FE-5680A units were cheap because they were considered end-of-life already.  Maybe for something somewhat fixable, like the 60 MHz VCXO center frequency having drifted out of a spec window.

> From: Chris Albertson <albertson.chris@gmail.com> > I'm ready to mount on of my 5680A in a case and looking for suggestions on > what one would consider a reasonable maximum case temp. With the FE-5680A mounted to a 4x6x1" finned aluminum heatsink as shown here: https://picasaweb.google.com/109928236040342205185/FE5680A#5685636673930493138 the opposite side of the case reaches 55 C in a 20 C ambient, with no fan whatsoever. Running a 40mm 12V fan at 5V aimed at the heatsink drops that to 45 C. At 5V the fan is nearly silent, and the airflow is quite weak. Despite the internal temperature compensation via DDS offset updates that the unit apparently does, I measured one of my units to have a 7E-12 per degree C temperature sensitivity, larger than I expected. So I do want to use a thermostat on the fan. The Symmetricom XPRO manual http://www.symmetricom.com/link.cfm?lid=7082 (which I believe is a somewhat similar Rb design) estimates that the lifetime of that unit doubles when baseplate temperature is reduced from 60 C to 40 C (MTBF: 243k hours => 591k hours). Note that 243k hours is about 28 years. On the other hand, my assumption is these surplus FE-5680A units were cheap because they were considered end-of-life already. Maybe for something somewhat fixable, like the 60 MHz VCXO center frequency having drifted out of a spec window.
PL
Pete Lancashire
Tue, Feb 28, 2012 9:30 PM

bad english on the 'if that changes'. Eventually with both in the same
box have the
Rb tweeked by the GPS. Just some other project for fun.

On Tue, Feb 28, 2012 at 11:42 AM, Chris Albertson
albertson.chris@gmail.com wrote:

On Tue, Feb 28, 2012 at 11:24 AM, Pete Lancashire
pete@petelancashire.com wrote:

I'm ready to mount on of my 5680A in a case and looking for suggestions on
what one would consider a reasonable maximum case temp. Not being
a blood sucking cell provider,  I can't replace a dead one every year.

I think in practice you are going to need to mount the 5680 to an
aluminum plate at least twice the size of the 5680 and then keep a
moderate airflow over the top and bottom.  Perhaps mount the  plate
using standoffs.   It's clear to me from seeing all those screws that
the 5680 needs to be firmly connected to some kind of heat sink even
it only a large plate.

Eventually I see locking the 5680A to GPS if that changes what one would

Yes you pretty much have to calibrate the 5680.   But what do you mean
"if that changes"?  the 5680 will drift away from any calibration.

Chris Albertson
Redondo Beach, California


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

bad english on the 'if that changes'. Eventually with both in the same box have the Rb tweeked by the GPS. Just some other project for fun. On Tue, Feb 28, 2012 at 11:42 AM, Chris Albertson <albertson.chris@gmail.com> wrote: > On Tue, Feb 28, 2012 at 11:24 AM, Pete Lancashire > <pete@petelancashire.com> wrote: >> I'm ready to mount on of my 5680A in a case and looking for suggestions on >> what one would consider a reasonable maximum case temp. Not being >> a blood sucking cell provider,  I can't replace a dead one every year. > > I think in practice you are going to need to mount the 5680 to an > aluminum plate at least twice the size of the 5680 and then keep a > moderate airflow over the top and bottom.  Perhaps mount the  plate > using standoffs.   It's clear to me from seeing all those screws that > the 5680 needs to be firmly connected to some kind of heat sink even > it only a large plate. >> >> Eventually I see locking the 5680A to GPS if that changes what one would > > Yes you pretty much have to calibrate the 5680.   But what do you mean > "if that changes"?  the 5680 will drift away from any calibration. > > > Chris Albertson > Redondo Beach, California > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Tue, Feb 28, 2012 10:13 PM

Hi

I would aim for a 40C base plate temperature under normal conditions. Other
parts of the unit will be a bit hotter. The temperature rise to the stuff
inside will be a few degrees. Not to hard to do with a modest fan.

Bob

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Pete Lancashire
Sent: Tuesday, February 28, 2012 2:25 PM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] 5680A what is a reasonable case temp ?

I'm ready to mount on of my 5680A in a case and looking for suggestions on
what one would consider a reasonable maximum case temp. Not being
a blood sucking cell provider,  I can't replace a dead one every year.

Eventually I see locking the 5680A to GPS if that changes what one would
suggest.

-pete


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I would aim for a 40C base plate temperature under normal conditions. Other parts of the unit will be a bit hotter. The temperature rise to the stuff inside will be a few degrees. Not to hard to do with a modest fan. Bob -----Original Message----- From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On Behalf Of Pete Lancashire Sent: Tuesday, February 28, 2012 2:25 PM To: Discussion of precise time and frequency measurement Subject: [time-nuts] 5680A what is a reasonable case temp ? I'm ready to mount on of my 5680A in a case and looking for suggestions on what one would consider a reasonable maximum case temp. Not being a blood sucking cell provider, I can't replace a dead one every year. Eventually I see locking the 5680A to GPS if that changes what one would suggest. -pete _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
TK
Tom Knox
Tue, Feb 28, 2012 10:22 PM

Hi Pete;
I saw a paper from on of the "Blood Suckers" that said the cooler the better, and the effect of temp on life was huge. I looked on my computer and couldn't find the paper again. I now think that data may have been in the hard copy of a manual for one of the telecom shelves I have worked with. If I find it, I will pass the numbers on. But needless to say, for long life keep it as cool as reasonably possible.
Best Wishes;
Thomas Knox

From: lists@rtty.us
To: time-nuts@febo.com
Date: Tue, 28 Feb 2012 17:13:12 -0500
Subject: Re: [time-nuts] 5680A what is a reasonable case temp ?

Hi

I would aim for a 40C base plate temperature under normal conditions. Other
parts of the unit will be a bit hotter. The temperature rise to the stuff
inside will be a few degrees. Not to hard to do with a modest fan.

Bob

-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Pete Lancashire
Sent: Tuesday, February 28, 2012 2:25 PM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] 5680A what is a reasonable case temp ?

I'm ready to mount on of my 5680A in a case and looking for suggestions on
what one would consider a reasonable maximum case temp. Not being
a blood sucking cell provider,  I can't replace a dead one every year.

Eventually I see locking the 5680A to GPS if that changes what one would
suggest.

-pete


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Pete; I saw a paper from on of the "Blood Suckers" that said the cooler the better, and the effect of temp on life was huge. I looked on my computer and couldn't find the paper again. I now think that data may have been in the hard copy of a manual for one of the telecom shelves I have worked with. If I find it, I will pass the numbers on. But needless to say, for long life keep it as cool as reasonably possible. Best Wishes; Thomas Knox > From: lists@rtty.us > To: time-nuts@febo.com > Date: Tue, 28 Feb 2012 17:13:12 -0500 > Subject: Re: [time-nuts] 5680A what is a reasonable case temp ? > > Hi > > I would aim for a 40C base plate temperature under normal conditions. Other > parts of the unit will be a bit hotter. The temperature rise to the stuff > inside will be a few degrees. Not to hard to do with a modest fan. > > Bob > > -----Original Message----- > From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On > Behalf Of Pete Lancashire > Sent: Tuesday, February 28, 2012 2:25 PM > To: Discussion of precise time and frequency measurement > Subject: [time-nuts] 5680A what is a reasonable case temp ? > > I'm ready to mount on of my 5680A in a case and looking for suggestions on > what one would consider a reasonable maximum case temp. Not being > a blood sucking cell provider, I can't replace a dead one every year. > > Eventually I see locking the 5680A to GPS if that changes what one would > suggest. > > -pete > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
AK
Attila Kinali
Wed, Feb 29, 2012 11:03 AM

On Tue, 28 Feb 2012 11:42:51 -0800
Chris Albertson albertson.chris@gmail.com wrote:

It's clear to me from seeing all those screws that
the 5680 needs to be firmly connected to some kind of heat sink even
it only a large plate.

It could also be that those screws are used to hold the top
and bottom part of the cover firmly together. Dont forget that
Rb's are sensitive to magnetic fields and that the case is mostlikely
made of some high permeability material.

		Attila Kinali

--
The trouble with you, Shev, is you don't say anything until you've saved
up a whole truckload of damned heavy brick arguments and then you dump
them all out and never look at the bleeding body mangled beneath the heap
-- Tirin, The Dispossessed, U. Le Guin

On Tue, 28 Feb 2012 11:42:51 -0800 Chris Albertson <albertson.chris@gmail.com> wrote: > It's clear to me from seeing all those screws that > the 5680 needs to be firmly connected to some kind of heat sink even > it only a large plate. It could also be that those screws are used to hold the top and bottom part of the cover firmly together. Dont forget that Rb's are sensitive to magnetic fields and that the case is mostlikely made of some high permeability material. Attila Kinali -- The trouble with you, Shev, is you don't say anything until you've saved up a whole truckload of damned heavy brick arguments and then you dump them all out and never look at the bleeding body mangled beneath the heap -- Tirin, The Dispossessed, U. Le Guin
PG
Peter Gottlieb
Wed, Feb 29, 2012 12:15 PM

I am fitting one of these into a HP 5335A counter.  The project became a bit
more involved than it first seemed and part of that is how to deal with the
cooling.  The manual shows an AC fan but it had been replaced (sloppily, I might
add) with a DC fan which was terribly noisy.  I found a much smoother one with
even more airflow in my junkbox.  The 5680A is mounted near where the fan is so
it gets plenty of airflow, but what about when the unit is in standby?  What I
did was put a little relay in which is powered by one of the DC supply outputs.
When the unit is off, the relay opens which places a resistor in series with the
fan (which is now powered by the small laptop 15 volt supply I installed).
Thus, when in standby the fan drops to a lower speed and acceptably quiet
operation and the 5680A doesn't get too hot.  I determined the resistor value
experimentally via my super duper accurate
run-unit-for-a-while-remove-cover-feel-5680A-case-change-resistor-repeat
method.  Ok so I'm not looking for supreme accuracy of temperature, just good
operation.

Nifty little project to get great accuracy for half the price of a good used HP
oven oscillator.

Peter

On 2/29/2012 6:03 AM, Attila Kinali wrote:

On Tue, 28 Feb 2012 11:42:51 -0800
Chris Albertsonalbertson.chris@gmail.com  wrote:

It's clear to me from seeing all those screws that
the 5680 needs to be firmly connected to some kind of heat sink even
it only a large plate.

It could also be that those screws are used to hold the top
and bottom part of the cover firmly together. Dont forget that
Rb's are sensitive to magnetic fields and that the case is mostlikely
made of some high permeability material.

		Attila Kinali
I am fitting one of these into a HP 5335A counter. The project became a bit more involved than it first seemed and part of that is how to deal with the cooling. The manual shows an AC fan but it had been replaced (sloppily, I might add) with a DC fan which was terribly noisy. I found a much smoother one with even more airflow in my junkbox. The 5680A is mounted near where the fan is so it gets plenty of airflow, but what about when the unit is in standby? What I did was put a little relay in which is powered by one of the DC supply outputs. When the unit is off, the relay opens which places a resistor in series with the fan (which is now powered by the small laptop 15 volt supply I installed). Thus, when in standby the fan drops to a lower speed and acceptably quiet operation and the 5680A doesn't get too hot. I determined the resistor value experimentally via my super duper accurate run-unit-for-a-while-remove-cover-feel-5680A-case-change-resistor-repeat method. Ok so I'm not looking for supreme accuracy of temperature, just good operation. Nifty little project to get great accuracy for half the price of a good used HP oven oscillator. Peter On 2/29/2012 6:03 AM, Attila Kinali wrote: > On Tue, 28 Feb 2012 11:42:51 -0800 > Chris Albertson<albertson.chris@gmail.com> wrote: > >> It's clear to me from seeing all those screws that >> the 5680 needs to be firmly connected to some kind of heat sink even >> it only a large plate. > It could also be that those screws are used to hold the top > and bottom part of the cover firmly together. Dont forget that > Rb's are sensitive to magnetic fields and that the case is mostlikely > made of some high permeability material. > > Attila Kinali >
CA
Chris Albertson
Wed, Feb 29, 2012 6:15 PM

That is one of the better uses for a 5680.  I have a counter that is
non-function because some one salvd the HP OCXO from it and I'm
thinking that a 5680 would fit inside.  But then I thought again and
the 5680 can fit outside.

One problem I think with your setup is that the 5680' temperature will
change abruptly when you take the counter off standby.  You might
have to wait 15 or 20 minutes for it to be stable.    It think it is
easy to just run the fan at the exact same minimum rate all the time
I put a temperature sensor in the 5680 and use a comparitor to turn
the fan off and on.  The 5680 needs 15 volts so I dropped some volts
with an LM317 regulator  I adjust the regulator voltage so the fan
does not make to much noise and the lm311 swiches the fan on/off as
required.  So there is never an abrupt change.  and the fan run no
more than it absolutely needs to.  I had to add a positive feedback
resistor to reduce the cycle rate.

On Wed, Feb 29, 2012 at 4:15 AM, Peter Gottlieb nerd@verizon.net wrote:

I am fitting one of these into a HP 5335A counter.  The project became a bit
more involved than it first seemed and part of that is how to deal with the
cooling.  The manual shows an AC fan but it had been replaced (sloppily, I
might add) with a DC fan which was terribly noisy.  I found a much smoother
one with even more airflow in my junkbox.  The 5680A is mounted near where
the fan is so it gets plenty of airflow, but what about when the unit is in
standby?  What I did was put a little relay in which is powered by one of
the DC supply outputs.  When the unit is off, the relay opens which places a
resistor in series with the fan (which is now powered by the small laptop 15
volt supply I installed).  Thus, when in standby the fan drops to a lower
speed and acceptably quiet operation and the 5680A doesn't get too hot.  I
determined the resistor value experimentally via my super duper accurate
run-unit-for-a-while-remove-cover-feel-5680A-case-change-resistor-repeat
method.  Ok so I'm not looking for supreme accuracy of temperature, just
good operation.

Nifty little project to get great accuracy for half the price of a good used
HP oven oscillator.

Peter

On 2/29/2012 6:03 AM, Attila Kinali wrote:

On Tue, 28 Feb 2012 11:42:51 -0800
Chris Albertsonalbertson.chris@gmail.com  wrote:

 It's clear to me from seeing all those screws that
the 5680 needs to be firmly connected to some kind of heat sink even
it only a large plate.

It could also be that those screws are used to hold the top
and bottom part of the cover firmly together. Dont forget that
Rb's are sensitive to magnetic fields and that the case is mostlikely
made of some high permeability material.

                       Attila Kinali


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--

Chris Albertson
Redondo Beach, California

That is one of the better uses for a 5680. I have a counter that is non-function because some one salvd the HP OCXO from it and I'm thinking that a 5680 would fit inside. But then I thought again and the 5680 can fit outside. One problem I think with your setup is that the 5680' temperature will change abruptly when you take the counter off standby. You might have to wait 15 or 20 minutes for it to be stable. It think it is easy to just run the fan at the exact same minimum rate all the time I put a temperature sensor in the 5680 and use a comparitor to turn the fan off and on. The 5680 needs 15 volts so I dropped some volts with an LM317 regulator I adjust the regulator voltage so the fan does not make to much noise and the lm311 swiches the fan on/off as required. So there is never an abrupt change. and the fan run no more than it absolutely needs to. I had to add a positive feedback resistor to reduce the cycle rate. On Wed, Feb 29, 2012 at 4:15 AM, Peter Gottlieb <nerd@verizon.net> wrote: > I am fitting one of these into a HP 5335A counter.  The project became a bit > more involved than it first seemed and part of that is how to deal with the > cooling.  The manual shows an AC fan but it had been replaced (sloppily, I > might add) with a DC fan which was terribly noisy.  I found a much smoother > one with even more airflow in my junkbox.  The 5680A is mounted near where > the fan is so it gets plenty of airflow, but what about when the unit is in > standby?  What I did was put a little relay in which is powered by one of > the DC supply outputs.  When the unit is off, the relay opens which places a > resistor in series with the fan (which is now powered by the small laptop 15 > volt supply I installed).  Thus, when in standby the fan drops to a lower > speed and acceptably quiet operation and the 5680A doesn't get too hot.  I > determined the resistor value experimentally via my super duper accurate > run-unit-for-a-while-remove-cover-feel-5680A-case-change-resistor-repeat > method.  Ok so I'm not looking for supreme accuracy of temperature, just > good operation. > > Nifty little project to get great accuracy for half the price of a good used > HP oven oscillator. > > Peter > > > > > > On 2/29/2012 6:03 AM, Attila Kinali wrote: >> >> On Tue, 28 Feb 2012 11:42:51 -0800 >> Chris Albertson<albertson.chris@gmail.com>  wrote: >> >>>  It's clear to me from seeing all those screws that >>> the 5680 needs to be firmly connected to some kind of heat sink even >>> it only a large plate. >> >> It could also be that those screws are used to hold the top >> and bottom part of the cover firmly together. Dont forget that >> Rb's are sensitive to magnetic fields and that the case is mostlikely >> made of some high permeability material. >> >>                        Attila Kinali >> > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. -- Chris Albertson Redondo Beach, California
BC
Bob Camp
Thu, Mar 1, 2012 12:24 AM

Hi

If it is a normal FE it is pretty close at 15 minutes and quite good after an hour or so. It certainly will beat the original OCXO at at hour. You might just run it when the counter is on and forget about running it in the standby mode.

Bob

On Feb 29, 2012, at 1:15 PM, Chris Albertson albertson.chris@gmail.com wrote:

That is one of the better uses for a 5680.  I have a counter that is
non-function because some one salvd the HP OCXO from it and I'm
thinking that a 5680 would fit inside.  But then I thought again and
the 5680 can fit outside.

One problem I think with your setup is that the 5680' temperature will
change abruptly when you take the counter off standby.  You might
have to wait 15 or 20 minutes for it to be stable.    It think it is
easy to just run the fan at the exact same minimum rate all the time
I put a temperature sensor in the 5680 and use a comparitor to turn
the fan off and on.  The 5680 needs 15 volts so I dropped some volts
with an LM317 regulator  I adjust the regulator voltage so the fan
does not make to much noise and the lm311 swiches the fan on/off as
required.  So there is never an abrupt change.  and the fan run no
more than it absolutely needs to.  I had to add a positive feedback
resistor to reduce the cycle rate.

On Wed, Feb 29, 2012 at 4:15 AM, Peter Gottlieb nerd@verizon.net wrote:

I am fitting one of these into a HP 5335A counter.  The project became a bit
more involved than it first seemed and part of that is how to deal with the
cooling.  The manual shows an AC fan but it had been replaced (sloppily, I
might add) with a DC fan which was terribly noisy.  I found a much smoother
one with even more airflow in my junkbox.  The 5680A is mounted near where
the fan is so it gets plenty of airflow, but what about when the unit is in
standby?  What I did was put a little relay in which is powered by one of
the DC supply outputs.  When the unit is off, the relay opens which places a
resistor in series with the fan (which is now powered by the small laptop 15
volt supply I installed).  Thus, when in standby the fan drops to a lower
speed and acceptably quiet operation and the 5680A doesn't get too hot.  I
determined the resistor value experimentally via my super duper accurate
run-unit-for-a-while-remove-cover-feel-5680A-case-change-resistor-repeat
method.  Ok so I'm not looking for supreme accuracy of temperature, just
good operation.

Nifty little project to get great accuracy for half the price of a good used
HP oven oscillator.

Peter

On 2/29/2012 6:03 AM, Attila Kinali wrote:

On Tue, 28 Feb 2012 11:42:51 -0800
Chris Albertsonalbertson.chris@gmail.com  wrote:

It's clear to me from seeing all those screws that
the 5680 needs to be firmly connected to some kind of heat sink even
it only a large plate.

It could also be that those screws are used to hold the top
and bottom part of the cover firmly together. Dont forget that
Rb's are sensitive to magnetic fields and that the case is mostlikely
made of some high permeability material.

                    Attila Kinali

time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--

Chris Albertson
Redondo Beach, California


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi If it is a normal FE it is pretty close at 15 minutes and quite good after an hour or so. It certainly will beat the original OCXO at at hour. You might just run it when the counter is on and forget about running it in the standby mode. Bob On Feb 29, 2012, at 1:15 PM, Chris Albertson <albertson.chris@gmail.com> wrote: > That is one of the better uses for a 5680. I have a counter that is > non-function because some one salvd the HP OCXO from it and I'm > thinking that a 5680 would fit inside. But then I thought again and > the 5680 can fit outside. > > One problem I think with your setup is that the 5680' temperature will > change abruptly when you take the counter off standby. You might > have to wait 15 or 20 minutes for it to be stable. It think it is > easy to just run the fan at the exact same minimum rate all the time > I put a temperature sensor in the 5680 and use a comparitor to turn > the fan off and on. The 5680 needs 15 volts so I dropped some volts > with an LM317 regulator I adjust the regulator voltage so the fan > does not make to much noise and the lm311 swiches the fan on/off as > required. So there is never an abrupt change. and the fan run no > more than it absolutely needs to. I had to add a positive feedback > resistor to reduce the cycle rate. > > On Wed, Feb 29, 2012 at 4:15 AM, Peter Gottlieb <nerd@verizon.net> wrote: >> I am fitting one of these into a HP 5335A counter. The project became a bit >> more involved than it first seemed and part of that is how to deal with the >> cooling. The manual shows an AC fan but it had been replaced (sloppily, I >> might add) with a DC fan which was terribly noisy. I found a much smoother >> one with even more airflow in my junkbox. The 5680A is mounted near where >> the fan is so it gets plenty of airflow, but what about when the unit is in >> standby? What I did was put a little relay in which is powered by one of >> the DC supply outputs. When the unit is off, the relay opens which places a >> resistor in series with the fan (which is now powered by the small laptop 15 >> volt supply I installed). Thus, when in standby the fan drops to a lower >> speed and acceptably quiet operation and the 5680A doesn't get too hot. I >> determined the resistor value experimentally via my super duper accurate >> run-unit-for-a-while-remove-cover-feel-5680A-case-change-resistor-repeat >> method. Ok so I'm not looking for supreme accuracy of temperature, just >> good operation. >> >> Nifty little project to get great accuracy for half the price of a good used >> HP oven oscillator. >> >> Peter >> >> >> >> >> >> On 2/29/2012 6:03 AM, Attila Kinali wrote: >>> >>> On Tue, 28 Feb 2012 11:42:51 -0800 >>> Chris Albertson<albertson.chris@gmail.com> wrote: >>> >>>> It's clear to me from seeing all those screws that >>>> the 5680 needs to be firmly connected to some kind of heat sink even >>>> it only a large plate. >>> >>> It could also be that those screws are used to hold the top >>> and bottom part of the cover firmly together. Dont forget that >>> Rb's are sensitive to magnetic fields and that the case is mostlikely >>> made of some high permeability material. >>> >>> Attila Kinali >>> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > > > -- > > Chris Albertson > Redondo Beach, California > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.