I would measure the counter both ways to characterize it, and you may
learn a little more about how it reacts.
Use its internal oscillator for the counter, apply the same 1PPS signal
to both inputs.
Use an external time base for the counter, apply the same 1 PPS to both
inputs - make sure the timebase is not related to the 1 PPS. Then
repeat, and use a common oscillator for all.
You may also introduce delay into the stop channel and see what that
causes and any differences, etc.
Log your results in your time-nuts log book, you may have a reason to
look at it again, down the road.
Brian
KD4FM
On Tue, 06 Nov 2012 02:14:04 +0100, Magnus Danielson
magnus@rubidium.dyndns.org wrote:
Hi Bob,
On 11/05/2012 01:30 PM, Bob Camp wrote:
Hi
As a practical example - a SR620 will look much better reading it's own reference than it will looking at almost anything else. That said, it's still a good idea to make sure the counter looks good reading it's own reference. If it doesn't look good, then you need to fix something.
I agree that this is a good self-test strategy, it usually gives a good
clue if something is really bad or not, but as I pointed out, it doesn't
give a fair idea of the measurement noise floor.
Cheers,
Magnus
A lot of counters support a self check mode which counts the internal
reference.
I tracked down an interesting problem in a Tektronix DC505 by wiring
the internal channel A input to measure the internal reference. Noise
from the display multiplexing was getting into the level shifter
located before the last 4 integrated counter stages and causing
spurious counts. Based on the documentation and how it did not match
the actual circuit, I suspect Tektronix fiddled with the values in
production until it seemed to work but never figured out the real
problem or back annotated the documentation.