Steve,

Flight Safety?  I got my Private Pilot's license in 1968, an ATP in 1992,
and now about 4500 hours later I am still flying.

I did find a 3458A that should be here next week.  It was advertised as
fully functional and comes with a 30 day return guarantee.  For what I paid,
I will be able to return it to Agilent for repair and still be ahead of what
I have seen on theBay.  Hopefully, all it will need is a return for
alignment/calibration.

If Solartron is 'on schedule', I should have the 7081 back before
Thanksgiving.  I just hope there is not hidden disaster that makes it
useless.

I got the paper manuals today for the 7081 and they appear to be in good
condition but I am a little suspicious that they are copies rather than
original.  But very good condition and very legible.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Steve Krull
Sent: Thursday, October 06, 2011 8:39 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron

Joe,

Yes, aviation it is. 34 years with one of the largest trainers of
corporate, commuter, private, and military pilots and mechanics, with
flight attendants thrown in for good measure. Unfortunately it involves
a lot of travel as I'm the interface with the aviation authorities
around the globe.

My 7081 is sitting on the bench warming up and I hope to do some
comparisons against the 3458A before I get called out of the country again.

I hope they provide a good report with the return of your meter. Let us
know how it turns out.

Best regards,

Steve

On 10/4/2011 9:53 PM, J. L. Trantham wrote:

Didier,

I will upload the 7061 manual later tonight.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]On
Behalf Of shalimr9@gmail.com
Sent: Wednesday, October 05, 2011 3:07 PM
To: Volt-Nuts
Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron

Joe,

I am interested in hosting your manuals on my web site.

Let me know if that's ok with you.

Didier KO4BB

Sent from my BlackBerry Wireless thingy while I do other things...

-----Original Message-----
From: "J. L. Trantham"jltran@att.net
Sender: volt-nuts-bounces@febo.com
Date: Sun, 25 Sep 2011 19:41:25
To: Discussion of precise voltage measurementvolt-nuts@febo.com
Reply-To: Discussion of precise voltage measurementvolt-nuts@febo.com
Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron

Steve,

You must have something to do with aviation.

My meter just arrived in Oak Ridge yesterday.  Hopefully, it will be on

the

way to England later this week.  Their turn around time estimate is 4

weeks.

Hopefully, I will have it back by first part of November.

I found the folks in Oak Ridge very friendly and helpful.  The lady I

spoke

with offered to try to find a manual for the 7081 but did not have any.

She

did, though, have an operating manual for the 7061 that she sent me in

.PDF

form.  If anyone wants it, let me know.  It is about 1 MB in size.  I have
not looked to see if one is posted anywhere around the net or not.

Also, I found a complete set of used manuals for the 7081 (Operating Part

I,

Operating Part II, and Service) at Manuals Plus and bought them.  They

have

yet to arrive.

I am anxious to see what they find in England and how it performs when it
gets back.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]On
Behalf Of Steve Krull
Sent: Tuesday, October 04, 2011 1:15 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron

Joe,

I was within a couple of hundred yards of your 7081 this past week, if
it shipped to England without any delay! The Ametek/Solartron building
is just up the road from our pilot training center at Farnborough
Airport and I was there for meetings this past week. I had intended to
wander up the road and see if there was anyone left from the test
equipment group but time got away from me. The closest I got was driving
by the entrance to the industrial park.

While in England I did manage, through a friend, to pick up a set of
Solartron test leads with banana plugs, and he's going to see if he can
find his Kelvin clip set and probe set too. With the leads, good food,
good ale, good weather, and good company, it was a great trip. I
generally have at least one meeting a year at Farnborough and Gatwick
airports so I hope to eventually make it up the road to
Ametek/Solartron. Not sure if I want to take my 7081 in my luggage but I
may yet.

Steve

On Sun, Sep 25, 2011 at 11:31 AM, J. L. Trantham wrote:

Fred,

I have recently acquired a Solartron 7081 and have had a lot of fun
getting
to know the meter.  In an effort to get it 'calibrated', I chased down
the
US branch of AMETEK, the company that now apparently owns Solartron.
I
shipped my meter to them yesterday and they will ship it to England
for
calibration, repair and alignment, as needed.  The 7081 ceased
production in
1998 and full support ended in 2008 but it is still supported, 'best
endeavours'.

When I was on the phone with them this AM to give them my credit card
information, I asked if they still supported the 7061, remembering
this
thread.  She said the 7061 is still on the list of supported
equipment.  So,
worst case scenario, you could contact AMETEK and see what options
might
exist for your 7061.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]On
Behalf Of Fred Schneider
Sent: Thursday, September 29, 2011 5:35 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and
solartron

The voltage collapses. I have to measure how much but not just a few
uV.
More like 200 mV if I remember it right. I know it is a resitive
divider but
the influence on each other is getting worse if it is on 24/7,
another sign
something is wrong. The 1V influences akso the 10 V so you keep it
adjusting. If i set 10 V correct and then the 1 V's the 10 V is off
again.

The 845 AB had its nicads removed and they placed a big zener instead.
The
most important thing, I made a regulated and good filtered
powersupply. I
placed the pcb for that in the space the nicds were mounted.
I also have a HP frequency standard that had a batterybox. But they
were
dead and i did not needed that so now i build a trimble thunderbolt
into
that space and used the 100 KHz output of the HP for the 10 MHz. That
is the
advantage of having gear with a battery option ;-)

I can think i can fix the solartron Problem if I can find a service
manual.
But now I can not even check the powersupply voltages because I do not
know
them. Filtercaps are good.

Fred

Fred,

What happens to the 10 volt position of the Fluke 731 when you attach
a 10

MegOhm input Voltmeter ?

I am sure you know that the 1 volt positions are just resistive
dividers

and any loading of any kind will drop their value.

Also, could you give details on what modifications you made to the
845AB ?

thanks,

Bill....WB6BNQ


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1410 / Virus Database: 1520/3922 - Release Date:
09/27/11


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Steve, Flight Safety? I got my Private Pilot's license in 1968, an ATP in 1992, and now about 4500 hours later I am still flying. I did find a 3458A that should be here next week. It was advertised as fully functional and comes with a 30 day return guarantee. For what I paid, I will be able to return it to Agilent for repair and still be ahead of what I have seen on theBay. Hopefully, all it will need is a return for alignment/calibration. If Solartron is 'on schedule', I should have the 7081 back before Thanksgiving. I just hope there is not hidden disaster that makes it useless. I got the paper manuals today for the 7081 and they appear to be in good condition but I am a little suspicious that they are copies rather than original. But very good condition and very legible. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Steve Krull Sent: Thursday, October 06, 2011 8:39 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron Joe, Yes, aviation it is. 34 years with one of the largest trainers of corporate, commuter, private, and military pilots and mechanics, with flight attendants thrown in for good measure. Unfortunately it involves a lot of travel as I'm the interface with the aviation authorities around the globe. My 7081 is sitting on the bench warming up and I hope to do some comparisons against the 3458A before I get called out of the country again. I hope they provide a good report with the return of your meter. Let us know how it turns out. Best regards, Steve On 10/4/2011 9:53 PM, J. L. Trantham wrote: > Didier, > > I will upload the 7061 manual later tonight. > > Joe > > -----Original Message----- > From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]On > Behalf Of shalimr9@gmail.com > Sent: Wednesday, October 05, 2011 3:07 PM > To: Volt-Nuts > Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron > > Joe, > > I am interested in hosting your manuals on my web site. > > Let me know if that's ok with you. > > Didier KO4BB > > Sent from my BlackBerry Wireless thingy while I do other things... > > -----Original Message----- > From: "J. L. Trantham"<jltran@att.net> > Sender: volt-nuts-bounces@febo.com > Date: Sun, 25 Sep 2011 19:41:25 > To: Discussion of precise voltage measurement<volt-nuts@febo.com> > Reply-To: Discussion of precise voltage measurement<volt-nuts@febo.com> > Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron > > Steve, > > You must have something to do with aviation. > > My meter just arrived in Oak Ridge yesterday. Hopefully, it will be on the > way to England later this week. Their turn around time estimate is 4 weeks. > Hopefully, I will have it back by first part of November. > > I found the folks in Oak Ridge very friendly and helpful. The lady I spoke > with offered to try to find a manual for the 7081 but did not have any. She > did, though, have an operating manual for the 7061 that she sent me in .PDF > form. If anyone wants it, let me know. It is about 1 MB in size. I have > not looked to see if one is posted anywhere around the net or not. > > Also, I found a complete set of used manuals for the 7081 (Operating Part I, > Operating Part II, and Service) at Manuals Plus and bought them. They have > yet to arrive. > > I am anxious to see what they find in England and how it performs when it > gets back. > > Joe > > -----Original Message----- > From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]On > Behalf Of Steve Krull > Sent: Tuesday, October 04, 2011 1:15 PM > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron > > > Joe, > > I was within a couple of hundred yards of your 7081 this past week, if > it shipped to England without any delay! The Ametek/Solartron building > is just up the road from our pilot training center at Farnborough > Airport and I was there for meetings this past week. I had intended to > wander up the road and see if there was anyone left from the test > equipment group but time got away from me. The closest I got was driving > by the entrance to the industrial park. > > While in England I did manage, through a friend, to pick up a set of > Solartron test leads with banana plugs, and he's going to see if he can > find his Kelvin clip set and probe set too. With the leads, good food, > good ale, good weather, and good company, it was a great trip. I > generally have at least one meeting a year at Farnborough and Gatwick > airports so I hope to eventually make it up the road to > Ametek/Solartron. Not sure if I want to take my 7081 in my luggage but I > may yet. > > Steve > > > > > On Sun, Sep 25, 2011 at 11:31 AM, J. L. Trantham wrote: > >> Fred, >> >> I have recently acquired a Solartron 7081 and have had a lot of fun >> getting >> to know the meter. In an effort to get it 'calibrated', I chased down >> the >> US branch of AMETEK, the company that now apparently owns Solartron. >> I >> shipped my meter to them yesterday and they will ship it to England >> for >> calibration, repair and alignment, as needed. The 7081 ceased >> production in >> 1998 and full support ended in 2008 but it is still supported, 'best >> endeavours'. >> >> When I was on the phone with them this AM to give them my credit card >> information, I asked if they still supported the 7061, remembering >> this >> thread. She said the 7061 is still on the list of supported >> equipment. So, >> worst case scenario, you could contact AMETEK and see what options >> might >> exist for your 7061. >> >> Joe >> >> -----Original Message----- >> From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com]On >> Behalf Of Fred Schneider >> Sent: Thursday, September 29, 2011 5:35 PM >> To: Discussion of precise voltage measurement >> Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and >> solartron >> >> The voltage collapses. I have to measure how much but not just a few >> uV. >> More like 200 mV if I remember it right. I know it is a resitive >> divider but >> the influence on each other is getting worse if it is on 24/7, >> another sign >> something is wrong. The 1V influences akso the 10 V so you keep it >> adjusting. If i set 10 V correct and then the 1 V's the 10 V is off >> again. >> >> >> The 845 AB had its nicads removed and they placed a big zener instead. >> The >> most important thing, I made a regulated and good filtered >> powersupply. I >> placed the pcb for that in the space the nicds were mounted. >> I also have a HP frequency standard that had a batterybox. But they >> were >> dead and i did not needed that so now i build a trimble thunderbolt >> into >> that space and used the 100 KHz output of the HP for the 10 MHz. That >> is the >> advantage of having gear with a battery option ;-) >> >> I can think i can fix the solartron Problem if I can find a service >> manual. >> But now I can not even check the powersupply voltages because I do not >> know >> them. Filtercaps are good. >> >> Fred >> >> >>> Fred, >>> >>> What happens to the 10 volt position of the Fluke 731 when you attach >>> a 10 >> MegOhm input Voltmeter ? >>> I am sure you know that the 1 volt positions are just resistive >>> dividers >> and any loading of any kind will drop their value. >>> Also, could you give details on what modifications you made to the >>> 845AB ? >>> >>> thanks, >>> >>> Bill....WB6BNQ >>> >>> >>> >>> _______________________________________________ >>> volt-nuts mailing list -- volt-nuts@febo.com >>> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >>> and follow the instructions there. >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. >> ----- >> No virus found in this message. >> Checked by AVG - www.avg.com >> Version: 10.0.1410 / Virus Database: 1520/3922 - Release Date: >> 09/27/11 >> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Михаил
Fri, Oct 7, 2011 4:24 AM

Joe,

Is your paper manual differ from the PDF-versions in the web? Are they contain
the Service Manual (not Maintenance Manual)?

Mickle T.

Friday, October 7, 2011, 7:31:36 AM, you wrote:
JLT> Steve,

JLT> Flight Safety?  I got my Private Pilot's license in 1968, an ATP in 1992,
JLT> and now about 4500 hours later I am still flying.

JLT> I did find a 3458A that should be here next week.  It was advertised as
JLT> fully functional and comes with a 30 day return guarantee.  For what I paid,
JLT> I will be able to return it to Agilent for repair and still be ahead of what
JLT> I have seen on theBay.  Hopefully, all it will need is a return for
JLT> alignment/calibration.

JLT> If Solartron is 'on schedule', I should have the 7081 back before
JLT> Thanksgiving.  I just hope there is not hidden disaster that makes it
JLT> useless.

JLT> I got the paper manuals today for the 7081 and they appear to be in good
JLT> condition but I am a little suspicious that they are copies rather than
JLT> original.  But very good condition and very legible.

JLT> Joe

Joe, Is your paper manual differ from the PDF-versions in the web? Are they contain the Service Manual (not Maintenance Manual)? Mickle T. Friday, October 7, 2011, 7:31:36 AM, you wrote: JLT> Steve, JLT> Flight Safety? I got my Private Pilot's license in 1968, an ATP in 1992, JLT> and now about 4500 hours later I am still flying. JLT> I did find a 3458A that should be here next week. It was advertised as JLT> fully functional and comes with a 30 day return guarantee. For what I paid, JLT> I will be able to return it to Agilent for repair and still be ahead of what JLT> I have seen on theBay. Hopefully, all it will need is a return for JLT> alignment/calibration. JLT> If Solartron is 'on schedule', I should have the 7081 back before JLT> Thanksgiving. I just hope there is not hidden disaster that makes it JLT> useless. JLT> I got the paper manuals today for the 7081 and they appear to be in good JLT> condition but I am a little suspicious that they are copies rather than JLT> original. But very good condition and very legible. JLT> Joe
JL
J. L. Trantham
Fri, Oct 7, 2011 12:55 PM

Mickle,

I have not had a chance to compare it page by page but it is the Maintenance
Manual, part number 70810014, Issue 5: August 1986, just like the .PDF copy
on the web.  Only it is much more legible.

The Operating Manual Part I is part number 70810010, like the .PDF on the
web, but it is a different issue, Issue 7: July 1994.  Again, I have not
compared it page by page but it looks the same.

The Operating Manual Part II is part number 70810011, Issue 5: June 1985,
like the .PDF on the web.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Ìèõàèë
Sent: Thursday, October 06, 2011 11:25 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron

Joe,

Is your paper manual differ from the PDF-versions in the web? Are they
contain
the Service Manual (not Maintenance Manual)?

Mickle T.

Friday, October 7, 2011, 7:31:36 AM, you wrote:
JLT> Steve,

JLT> Flight Safety?  I got my Private Pilot's license in 1968, an ATP in
1992,
JLT> and now about 4500 hours later I am still flying.

JLT> I did find a 3458A that should be here next week.  It was advertised as
JLT> fully functional and comes with a 30 day return guarantee.  For what I
paid,
JLT> I will be able to return it to Agilent for repair and still be ahead of
what
JLT> I have seen on theBay.  Hopefully, all it will need is a return for
JLT> alignment/calibration.

JLT> If Solartron is 'on schedule', I should have the 7081 back before
JLT> Thanksgiving.  I just hope there is not hidden disaster that makes it
JLT> useless.

JLT> I got the paper manuals today for the 7081 and they appear to be in
good
JLT> condition but I am a little suspicious that they are copies rather than
JLT> original.  But very good condition and very legible.

JLT> Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Mickle, I have not had a chance to compare it page by page but it is the Maintenance Manual, part number 70810014, Issue 5: August 1986, just like the .PDF copy on the web. Only it is much more legible. The Operating Manual Part I is part number 70810010, like the .PDF on the web, but it is a different issue, Issue 7: July 1994. Again, I have not compared it page by page but it looks the same. The Operating Manual Part II is part number 70810011, Issue 5: June 1985, like the .PDF on the web. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Ìèõàèë Sent: Thursday, October 06, 2011 11:25 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Serious problems fluke 731A, 720 and solartron Joe, Is your paper manual differ from the PDF-versions in the web? Are they contain the Service Manual (not Maintenance Manual)? Mickle T. Friday, October 7, 2011, 7:31:36 AM, you wrote: JLT> Steve, JLT> Flight Safety? I got my Private Pilot's license in 1968, an ATP in 1992, JLT> and now about 4500 hours later I am still flying. JLT> I did find a 3458A that should be here next week. It was advertised as JLT> fully functional and comes with a 30 day return guarantee. For what I paid, JLT> I will be able to return it to Agilent for repair and still be ahead of what JLT> I have seen on theBay. Hopefully, all it will need is a return for JLT> alignment/calibration. JLT> If Solartron is 'on schedule', I should have the 7081 back before JLT> Thanksgiving. I just hope there is not hidden disaster that makes it JLT> useless. JLT> I got the paper manuals today for the 7081 and they appear to be in good JLT> condition but I am a little suspicious that they are copies rather than JLT> original. But very good condition and very legible. JLT> Joe _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.