AK
Attila Kinali
Thu, Oct 22, 2020 5:13 PM
Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?
Maybe try to get hold of one of the Philips (later licensed to
Fluke) PM6680 or PM6681? These are more common in Europe than
in the US, so the big US dominated websites/forums/.. don't
mention them that often. Solid devices that can be had as low
as 300€ if you are willing to wait, 500-800€ is the usual going
price. The SR620 is the workhorse that drives a lot of the
time and frequency metrology worldwide and can be had new and
used (new on http://thinksrs.com goes for 800-2000€ used).
If you go for a new one, I would consider looking at the
Pendulum CNT-90 and CNT-91. (Pendulum is the company that
took over Philips frequency counter business and the CNT-90
is the continuation of the PM668x line, also sold as PM6690
by Fluke)
Attila Kinali
--
<JaberWorky> The bad part of Zurich is where the degenerates
throw DARK chocolate at you.
On Thu, 22 Oct 2020 11:50:08 +0000
Giorgio Barinetti <giorgio@barinetti.it> wrote:
> Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
> But again : which one between the 3 ? 53131, 53132 or 53181 ?
Maybe try to get hold of one of the Philips (later licensed to
Fluke) PM6680 or PM6681? These are more common in Europe than
in the US, so the big US dominated websites/forums/.. don't
mention them that often. Solid devices that can be had as low
as 300€ if you are willing to wait, 500-800€ is the usual going
price. The SR620 is the workhorse that drives a lot of the
time and frequency metrology worldwide and can be had new and
used (new on http://thinksrs.com goes for 800-2000€ used).
If you go for a new one, I would consider looking at the
Pendulum CNT-90 and CNT-91. (Pendulum is the company that
took over Philips frequency counter business and the CNT-90
is the continuation of the PM668x line, also sold as PM6690
by Fluke)
Attila Kinali
--
<JaberWorky> The bad part of Zurich is where the degenerates
throw DARK chocolate at you.
O
Oz-in-DFW
Thu, Oct 22, 2020 7:08 PM
On 10/22/2020 11:52 AM, jimlux wrote:
I wonder if one could implement the protocol using GPIO pins on
something like a Beagleboard or Arduino clone?
(yes, as a product, with the right connector and line
driver/receivers, etc. - it would cost the same as a Prologix... but
as a hack...)
After all GPIB is a very old standard, and was developed back in the
day of 1 MHz logic. I think you could do the handshaking in software -
punch the bits one by one, assert DAV, etc.
Listener reads bits one by one, then asserts the ACK or NAK
Actually it's bytes. It's byte-serial not bit serial.
Most modern micros can pretty easily keep up with the bus - especially
if you use GPIOs that can generate interrupts. The only real challenges
are finding a reliable source of the appropriate connector - and driving
the 5V lines at the current levels required for a loaded bus. Most
modern micros do not have 5V tolerant I/O but that's pretty easily dealt
with. Connectors can be had, NORCOMP's 112 series should be ideal.
https://www.norcomp.net/series/112-series and are available through
DigiKey, Mouser, Farnell, etc. TI's SN75160/SN75161 bus transceivers
still seem to be in production as SOICs as well.
I've noodled about building a board that was intended to connect to a
single instrument, but not really gotten serious about it. I wanted
everything in the shop on Ethernet.
Oz (in DFW)
On 10/22/2020 11:52 AM, jimlux wrote:
> I wonder if one could implement the protocol using GPIO pins on
> something like a Beagleboard or Arduino clone?
>
> (yes, as a product, with the right connector and line
> driver/receivers, etc. - it would cost the same as a Prologix... but
> as a hack...)
>
> After all GPIB is a very old standard, and was developed back in the
> day of 1 MHz logic. I think you could do the handshaking in software -
> punch the bits one by one, assert DAV, etc.
> Listener reads bits one by one, then asserts the ACK or NAK
Actually it's bytes. It's byte-serial not bit serial.
Most modern micros can pretty easily keep up with the bus - especially
if you use GPIOs that can generate interrupts. The only real challenges
are finding a reliable source of the appropriate connector - and driving
the 5V lines at the current levels required for a loaded bus. Most
modern micros do not have 5V tolerant I/O but that's pretty easily dealt
with. Connectors can be had, NORCOMP's 112 series should be ideal.
https://www.norcomp.net/series/112-series and are available through
DigiKey, Mouser, Farnell, etc. TI's SN75160/SN75161 bus transceivers
still seem to be in production as SOICs as well.
I've noodled about building a board that was intended to connect to a
single instrument, but not really gotten serious about it. I wanted
everything in the shop on Ethernet.
Oz (in DFW)
J
jimlux
Fri, Oct 23, 2020 12:53 PM
On 10/22/20 12:08 PM, Oz-in-DFW wrote:
On 10/22/2020 11:52 AM, jimlux wrote:
I wonder if one could implement the protocol using GPIO pins on
something like a Beagleboard or Arduino clone?
(yes, as a product, with the right connector and line
driver/receivers, etc. - it would cost the same as a Prologix... but
as a hack...)
After all GPIB is a very old standard, and was developed back in the
day of 1 MHz logic. I think you could do the handshaking in software -
punch the bits one by one, assert DAV, etc.
Listener reads bits one by one, then asserts the ACK or NAK
Actually it's bytes. It's byte-serial not bit serial.
Yes, but sometimes you don't have a way to write all the bits of an 8
bit interface at once, and you have to do them one at a time (if you're
going through some abstracted interface that only allows individual pins
to be set - for instance, Arduinos work that way:
digitalWrite(pin#, HIGH)
I think GPIB would still work if you had to do 8 digitalWrite() calls,
then a final digitalWrite() call to assert DAV.
I suspect that for a number of Arduino type processors, there is a way
to write or read all 8 at once, assuming you were clever enough to pick
the right pins to use.
Most modern micros can pretty easily keep up with the bus - especially
if you use GPIOs that can generate interrupts.
I think that since the bus is fully handshaked (handshook?) an
arbitrarily slow processor can do it, no interrupts required.
The only real challenges
are finding a reliable source of the appropriate connector - and driving
the 5V lines at the current levels required for a loaded bus. Most
modern micros do not have 5V tolerant I/O but that's pretty easily dealt
with. Connectors can be had, NORCOMP's 112 series should be ideal.
https://www.norcomp.net/series/112-series and are available through
DigiKey, Mouser, Farnell, etc. TI's SN75160/SN75161 bus transceivers
still seem to be in production as SOICs as well.
I've noodled about building a board that was intended to connect to a
single instrument, but not really gotten serious about it. I wanted
everything in the shop on Ethernet.
Buy a Prologix - $200 from sparkfun, for instance.
Unless your time is free and you don't have any more interesting
projects to work on.
I wonder if Prologix makes one that is PoE? Now that you can get cheap
switches with PoE injection, that would be attractive. Getting rid of
all those wall warts would be nice.
On 10/22/20 12:08 PM, Oz-in-DFW wrote:
> On 10/22/2020 11:52 AM, jimlux wrote:
>> I wonder if one could implement the protocol using GPIO pins on
>> something like a Beagleboard or Arduino clone?
>>
>> (yes, as a product, with the right connector and line
>> driver/receivers, etc. - it would cost the same as a Prologix... but
>> as a hack...)
>>
>> After all GPIB is a very old standard, and was developed back in the
>> day of 1 MHz logic. I think you could do the handshaking in software -
>> punch the bits one by one, assert DAV, etc.
>> Listener reads bits one by one, then asserts the ACK or NAK
> Actually it's bytes. It's byte-serial not bit serial.
Yes, but sometimes you don't have a way to write all the bits of an 8
bit interface at once, and you have to do them one at a time (if you're
going through some abstracted interface that only allows individual pins
to be set - for instance, Arduinos work that way:
digitalWrite(pin#, HIGH)
I think GPIB would still work if you had to do 8 digitalWrite() calls,
then a final digitalWrite() call to assert DAV.
I suspect that for a number of Arduino type processors, there is a way
to write or read all 8 at once, assuming you were clever enough to pick
the right pins to use.
>
> Most modern micros can pretty easily keep up with the bus - especially
> if you use GPIOs that can generate interrupts.
I think that since the bus is fully handshaked (handshook?) an
arbitrarily slow processor can do it, no interrupts required.
The only real challenges
> are finding a reliable source of the appropriate connector - and driving
> the 5V lines at the current levels required for a loaded bus. Most
> modern micros do not have 5V tolerant I/O but that's pretty easily dealt
> with. Connectors can be had, NORCOMP's 112 series should be ideal.
> https://www.norcomp.net/series/112-series and are available through
> DigiKey, Mouser, Farnell, etc. TI's SN75160/SN75161 bus transceivers
> still seem to be in production as SOICs as well.
>
> I've noodled about building a board that was intended to connect to a
> single instrument, but not really gotten serious about it. I wanted
> everything in the shop on Ethernet.
Buy a Prologix - $200 from sparkfun, for instance.
Unless your time is free and you don't have any more interesting
projects to work on.
I wonder if Prologix makes one that is PoE? Now that you can get cheap
switches with PoE injection, that would be attractive. Getting rid of
all those wall warts would be nice.
>
> Oz (in DFW)
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
>
MD
Magnus Danielson
Fri, Oct 23, 2020 3:25 PM
Hi,
On 2020-10-22 19:13, Attila Kinali wrote:
Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?
Maybe try to get hold of one of the Philips (later licensed to
Fluke) PM6680 or PM6681? These are more common in Europe than
in the US, so the big US dominated websites/forums/.. don't
mention them that often. Solid devices that can be had as low
as 300€ if you are willing to wait, 500-800€ is the usual going
price. The SR620 is the workhorse that drives a lot of the
time and frequency metrology worldwide and can be had new and
used (new on http://thinksrs.com goes for 800-2000€ used).
If you go for a new one, I would consider looking at the
Pendulum CNT-90 and CNT-91. (Pendulum is the company that
took over Philips frequency counter business and the CNT-90
is the continuation of the PM668x line, also sold as PM6690
by Fluke)
Let me correct on the history and geniology there.
Philips had a instrument making side called Philips Industrier Järfälla
that did a range of measurement instruments. Later they joined forces
with Fluke. Later Philips felt that the business unit was a bad fit to
stay in Philips, so they sold it off to become a separate company which
became Pendelum. Pendelum was really the business unit with people etc
through that process, and the Fluke relation and rebranding continued.
Naturally Pendelum moved out of the Philips Industrier Järfälla office
over to Bälstabro (both locations in north of Stockholm) as it was sold
off. Pendelum also managed to rebrand their counters to Tektronix, which
mainly consisted of cosmetic changes to get the look and feel. Pendelum
was operated for many years like this, some of their production in
Pajala, where as other where done in Bälstabro. Later they reshaped the
production so that it moved to Poland where it remains. Pendulum was
sold to Spectracom and was operated as a subsidary for a while, until
they shut operation down.
The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
had the timing ASIC. The CNT-90 (100 ps) was developed to the CNT-91 (50
ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
CNT-92, but could not at that time do it with the same technical setup.
They also had the Wander Meter WM-10 which aided in testing telecom
sync. After some testing, I suggested they would broaden the product to
handle more signals and that is when they mostly firmware upgraded it to
the WM-11, and first time I tested the WM-11 the second '1' was cut out
from another '10' print and put over the '0'. Also, it was still under
development. Several years later they where still going back to my list
of feedback. They later did a revamped this into a new instrument with
even improved capabilities such as interfaces boards. This was later
sold off to Calnex which sells it as Sentinel, which is a great product
for telecom operators.
I still have contact with some of the Philips/Fluke/Pendelum staff and
owner.
Cheers,
Magnus
Hi,
On 2020-10-22 19:13, Attila Kinali wrote:
> On Thu, 22 Oct 2020 11:50:08 +0000
> Giorgio Barinetti <giorgio@barinetti.it> wrote:
>
>> Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
>> But again : which one between the 3 ? 53131, 53132 or 53181 ?
> Maybe try to get hold of one of the Philips (later licensed to
> Fluke) PM6680 or PM6681? These are more common in Europe than
> in the US, so the big US dominated websites/forums/.. don't
> mention them that often. Solid devices that can be had as low
> as 300€ if you are willing to wait, 500-800€ is the usual going
> price. The SR620 is the workhorse that drives a lot of the
> time and frequency metrology worldwide and can be had new and
> used (new on http://thinksrs.com goes for 800-2000€ used).
>
> If you go for a new one, I would consider looking at the
> Pendulum CNT-90 and CNT-91. (Pendulum is the company that
> took over Philips frequency counter business and the CNT-90
> is the continuation of the PM668x line, also sold as PM6690
> by Fluke)
Let me correct on the history and geniology there.
Philips had a instrument making side called Philips Industrier Järfälla
that did a range of measurement instruments. Later they joined forces
with Fluke. Later Philips felt that the business unit was a bad fit to
stay in Philips, so they sold it off to become a separate company which
became Pendelum. Pendelum was really the business unit with people etc
through that process, and the Fluke relation and rebranding continued.
Naturally Pendelum moved out of the Philips Industrier Järfälla office
over to Bälstabro (both locations in north of Stockholm) as it was sold
off. Pendelum also managed to rebrand their counters to Tektronix, which
mainly consisted of cosmetic changes to get the look and feel. Pendelum
was operated for many years like this, some of their production in
Pajala, where as other where done in Bälstabro. Later they reshaped the
production so that it moved to Poland where it remains. Pendulum was
sold to Spectracom and was operated as a subsidary for a while, until
they shut operation down.
The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
had the timing ASIC. The CNT-90 (100 ps) was developed to the CNT-91 (50
ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
CNT-92, but could not at that time do it with the same technical setup.
They also had the Wander Meter WM-10 which aided in testing telecom
sync. After some testing, I suggested they would broaden the product to
handle more signals and that is when they mostly firmware upgraded it to
the WM-11, and first time I tested the WM-11 the second '1' was cut out
from another '10' print and put over the '0'. Also, it was still under
development. Several years later they where still going back to my list
of feedback. They later did a revamped this into a new instrument with
even improved capabilities such as interfaces boards. This was later
sold off to Calnex which sells it as Sentinel, which is a great product
for telecom operators.
I still have contact with some of the Philips/Fluke/Pendelum staff and
owner.
Cheers,
Magnus
AB
Azelio Boriani
Thu, Oct 29, 2020 1:17 PM
For those who have the PM6681 (aka CNT81): check the 3V memory backup
cell and replace it before the dreaded calibration lost (CAL LOST)
will appear on the LCD. Replace the coin cell with great care (with
the counter powered up), see the service manual for the procedure.
https://archive.org/details/FLUKE_PM6681_Service_Manual
On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson magnus@rubidium.se wrote:
Hi,
On 2020-10-22 19:13, Attila Kinali wrote:
Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?
Maybe try to get hold of one of the Philips (later licensed to
Fluke) PM6680 or PM6681? These are more common in Europe than
in the US, so the big US dominated websites/forums/.. don't
mention them that often. Solid devices that can be had as low
as 300€ if you are willing to wait, 500-800€ is the usual going
price. The SR620 is the workhorse that drives a lot of the
time and frequency metrology worldwide and can be had new and
used (new on http://thinksrs.com goes for 800-2000€ used).
If you go for a new one, I would consider looking at the
Pendulum CNT-90 and CNT-91. (Pendulum is the company that
took over Philips frequency counter business and the CNT-90
is the continuation of the PM668x line, also sold as PM6690
by Fluke)
Let me correct on the history and geniology there.
Philips had a instrument making side called Philips Industrier Järfälla
that did a range of measurement instruments. Later they joined forces
with Fluke. Later Philips felt that the business unit was a bad fit to
stay in Philips, so they sold it off to become a separate company which
became Pendelum. Pendelum was really the business unit with people etc
through that process, and the Fluke relation and rebranding continued.
Naturally Pendelum moved out of the Philips Industrier Järfälla office
over to Bälstabro (both locations in north of Stockholm) as it was sold
off. Pendelum also managed to rebrand their counters to Tektronix, which
mainly consisted of cosmetic changes to get the look and feel. Pendelum
was operated for many years like this, some of their production in
Pajala, where as other where done in Bälstabro. Later they reshaped the
production so that it moved to Poland where it remains. Pendulum was
sold to Spectracom and was operated as a subsidary for a while, until
they shut operation down.
The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
had the timing ASIC. The CNT-90 (100 ps) was developed to the CNT-91 (50
ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
CNT-92, but could not at that time do it with the same technical setup.
They also had the Wander Meter WM-10 which aided in testing telecom
sync. After some testing, I suggested they would broaden the product to
handle more signals and that is when they mostly firmware upgraded it to
the WM-11, and first time I tested the WM-11 the second '1' was cut out
from another '10' print and put over the '0'. Also, it was still under
development. Several years later they where still going back to my list
of feedback. They later did a revamped this into a new instrument with
even improved capabilities such as interfaces boards. This was later
sold off to Calnex which sells it as Sentinel, which is a great product
for telecom operators.
I still have contact with some of the Philips/Fluke/Pendelum staff and
owner.
Cheers,
Magnus
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
For those who have the PM6681 (aka CNT81): check the 3V memory backup
cell and replace it before the dreaded calibration lost (CAL LOST)
will appear on the LCD. Replace the coin cell with great care (with
the counter powered up), see the service manual for the procedure.
<https://archive.org/details/FLUKE_PM6681_Service_Manual>
On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson <magnus@rubidium.se> wrote:
>
> Hi,
>
> On 2020-10-22 19:13, Attila Kinali wrote:
> > On Thu, 22 Oct 2020 11:50:08 +0000
> > Giorgio Barinetti <giorgio@barinetti.it> wrote:
> >
> >> Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
> >> But again : which one between the 3 ? 53131, 53132 or 53181 ?
> > Maybe try to get hold of one of the Philips (later licensed to
> > Fluke) PM6680 or PM6681? These are more common in Europe than
> > in the US, so the big US dominated websites/forums/.. don't
> > mention them that often. Solid devices that can be had as low
> > as 300€ if you are willing to wait, 500-800€ is the usual going
> > price. The SR620 is the workhorse that drives a lot of the
> > time and frequency metrology worldwide and can be had new and
> > used (new on http://thinksrs.com goes for 800-2000€ used).
> >
> > If you go for a new one, I would consider looking at the
> > Pendulum CNT-90 and CNT-91. (Pendulum is the company that
> > took over Philips frequency counter business and the CNT-90
> > is the continuation of the PM668x line, also sold as PM6690
> > by Fluke)
>
> Let me correct on the history and geniology there.
>
> Philips had a instrument making side called Philips Industrier Järfälla
> that did a range of measurement instruments. Later they joined forces
> with Fluke. Later Philips felt that the business unit was a bad fit to
> stay in Philips, so they sold it off to become a separate company which
> became Pendelum. Pendelum was really the business unit with people etc
> through that process, and the Fluke relation and rebranding continued.
> Naturally Pendelum moved out of the Philips Industrier Järfälla office
> over to Bälstabro (both locations in north of Stockholm) as it was sold
> off. Pendelum also managed to rebrand their counters to Tektronix, which
> mainly consisted of cosmetic changes to get the look and feel. Pendelum
> was operated for many years like this, some of their production in
> Pajala, where as other where done in Bälstabro. Later they reshaped the
> production so that it moved to Poland where it remains. Pendulum was
> sold to Spectracom and was operated as a subsidary for a while, until
> they shut operation down.
>
> The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
> had the timing ASIC. The CNT-90 (100 ps) was developed to the CNT-91 (50
> ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
> CNT-92, but could not at that time do it with the same technical setup.
> They also had the Wander Meter WM-10 which aided in testing telecom
> sync. After some testing, I suggested they would broaden the product to
> handle more signals and that is when they mostly firmware upgraded it to
> the WM-11, and first time I tested the WM-11 the second '1' was cut out
> from another '10' print and put over the '0'. Also, it was still under
> development. Several years later they where still going back to my list
> of feedback. They later did a revamped this into a new instrument with
> even improved capabilities such as interfaces boards. This was later
> sold off to Calnex which sells it as Sentinel, which is a great product
> for telecom operators.
>
> I still have contact with some of the Philips/Fluke/Pendelum staff and
> owner.
>
> Cheers,
> Magnus
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
MD
Magnus Danielson
Fri, Oct 30, 2020 12:37 AM
Hi,
I second this. You can read the calibration data out of the counter and
save. I've done some experiments with that, but nothing conclusive, but
I blame my lack of patience and not a proper setup.
Do replace the battery, it is cheap and relatively easy to do.
Would you loose this calibration, through a little GPIB commands one can
write a fake value in. This will however not produce the best resu. The
calibration routine actually runs an off beat frequency and then test
different values, and look for least RMS value, because it is the
calibration point. I've not had time to replicate all that, but I did
manage to write the fake value in and at one time get rid of the CAL
LOST warning.
Cheers,
Magnus
On 2020-10-29 14:17, Azelio Boriani wrote:
For those who have the PM6681 (aka CNT81): check the 3V memory backup
cell and replace it before the dreaded calibration lost (CAL LOST)
will appear on the LCD. Replace the coin cell with great care (with
the counter powered up), see the service manual for the procedure.
https://archive.org/details/FLUKE_PM6681_Service_Manual
On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson magnus@rubidium.se wrote:
Hi,
On 2020-10-22 19:13, Attila Kinali wrote:
Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?
Maybe try to get hold of one of the Philips (later licensed to
Fluke) PM6680 or PM6681? These are more common in Europe than
in the US, so the big US dominated websites/forums/.. don't
mention them that often. Solid devices that can be had as low
as 300€ if you are willing to wait, 500-800€ is the usual going
price. The SR620 is the workhorse that drives a lot of the
time and frequency metrology worldwide and can be had new and
used (new on http://thinksrs.com goes for 800-2000€ used).
If you go for a new one, I would consider looking at the
Pendulum CNT-90 and CNT-91. (Pendulum is the company that
took over Philips frequency counter business and the CNT-90
is the continuation of the PM668x line, also sold as PM6690
by Fluke)
Let me correct on the history and geniology there.
Philips had a instrument making side called Philips Industrier Järfälla
that did a range of measurement instruments. Later they joined forces
with Fluke. Later Philips felt that the business unit was a bad fit to
stay in Philips, so they sold it off to become a separate company which
became Pendelum. Pendelum was really the business unit with people etc
through that process, and the Fluke relation and rebranding continued.
Naturally Pendelum moved out of the Philips Industrier Järfälla office
over to Bälstabro (both locations in north of Stockholm) as it was sold
off. Pendelum also managed to rebrand their counters to Tektronix, which
mainly consisted of cosmetic changes to get the look and feel. Pendelum
was operated for many years like this, some of their production in
Pajala, where as other where done in Bälstabro. Later they reshaped the
production so that it moved to Poland where it remains. Pendulum was
sold to Spectracom and was operated as a subsidary for a while, until
they shut operation down.
The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
had the timing ASIC. The CNT-90 (100 ps) was developed to the CNT-91 (50
ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
CNT-92, but could not at that time do it with the same technical setup.
They also had the Wander Meter WM-10 which aided in testing telecom
sync. After some testing, I suggested they would broaden the product to
handle more signals and that is when they mostly firmware upgraded it to
the WM-11, and first time I tested the WM-11 the second '1' was cut out
from another '10' print and put over the '0'. Also, it was still under
development. Several years later they where still going back to my list
of feedback. They later did a revamped this into a new instrument with
even improved capabilities such as interfaces boards. This was later
sold off to Calnex which sells it as Sentinel, which is a great product
for telecom operators.
I still have contact with some of the Philips/Fluke/Pendelum staff and
owner.
Cheers,
Magnus
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi,
I second this. You can read the calibration data out of the counter and
save. I've done some experiments with that, but nothing conclusive, but
I blame my lack of patience and not a proper setup.
Do replace the battery, it is cheap and relatively easy to do.
Would you loose this calibration, through a little GPIB commands one can
write a fake value in. This will however not produce the best resu. The
calibration routine actually runs an off beat frequency and then test
different values, and look for least RMS value, because it is the
calibration point. I've not had time to replicate all that, but I did
manage to write the fake value in and at one time get rid of the CAL
LOST warning.
Cheers,
Magnus
On 2020-10-29 14:17, Azelio Boriani wrote:
> For those who have the PM6681 (aka CNT81): check the 3V memory backup
> cell and replace it before the dreaded calibration lost (CAL LOST)
> will appear on the LCD. Replace the coin cell with great care (with
> the counter powered up), see the service manual for the procedure.
> <https://archive.org/details/FLUKE_PM6681_Service_Manual>
>
> On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson <magnus@rubidium.se> wrote:
>> Hi,
>>
>> On 2020-10-22 19:13, Attila Kinali wrote:
>>> On Thu, 22 Oct 2020 11:50:08 +0000
>>> Giorgio Barinetti <giorgio@barinetti.it> wrote:
>>>
>>>> Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
>>>> But again : which one between the 3 ? 53131, 53132 or 53181 ?
>>> Maybe try to get hold of one of the Philips (later licensed to
>>> Fluke) PM6680 or PM6681? These are more common in Europe than
>>> in the US, so the big US dominated websites/forums/.. don't
>>> mention them that often. Solid devices that can be had as low
>>> as 300€ if you are willing to wait, 500-800€ is the usual going
>>> price. The SR620 is the workhorse that drives a lot of the
>>> time and frequency metrology worldwide and can be had new and
>>> used (new on http://thinksrs.com goes for 800-2000€ used).
>>>
>>> If you go for a new one, I would consider looking at the
>>> Pendulum CNT-90 and CNT-91. (Pendulum is the company that
>>> took over Philips frequency counter business and the CNT-90
>>> is the continuation of the PM668x line, also sold as PM6690
>>> by Fluke)
>> Let me correct on the history and geniology there.
>>
>> Philips had a instrument making side called Philips Industrier Järfälla
>> that did a range of measurement instruments. Later they joined forces
>> with Fluke. Later Philips felt that the business unit was a bad fit to
>> stay in Philips, so they sold it off to become a separate company which
>> became Pendelum. Pendelum was really the business unit with people etc
>> through that process, and the Fluke relation and rebranding continued.
>> Naturally Pendelum moved out of the Philips Industrier Järfälla office
>> over to Bälstabro (both locations in north of Stockholm) as it was sold
>> off. Pendelum also managed to rebrand their counters to Tektronix, which
>> mainly consisted of cosmetic changes to get the look and feel. Pendelum
>> was operated for many years like this, some of their production in
>> Pajala, where as other where done in Bälstabro. Later they reshaped the
>> production so that it moved to Poland where it remains. Pendulum was
>> sold to Spectracom and was operated as a subsidary for a while, until
>> they shut operation down.
>>
>> The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
>> had the timing ASIC. The CNT-90 (100 ps) was developed to the CNT-91 (50
>> ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
>> CNT-92, but could not at that time do it with the same technical setup.
>> They also had the Wander Meter WM-10 which aided in testing telecom
>> sync. After some testing, I suggested they would broaden the product to
>> handle more signals and that is when they mostly firmware upgraded it to
>> the WM-11, and first time I tested the WM-11 the second '1' was cut out
>> from another '10' print and put over the '0'. Also, it was still under
>> development. Several years later they where still going back to my list
>> of feedback. They later did a revamped this into a new instrument with
>> even improved capabilities such as interfaces boards. This was later
>> sold off to Calnex which sells it as Sentinel, which is a great product
>> for telecom operators.
>>
>> I still have contact with some of the Philips/Fluke/Pendelum staff and
>> owner.
>>
>> Cheers,
>> Magnus
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
DJ
Didier Juges
Fri, Oct 30, 2020 12:50 PM
Just for reference, I have an HP5334 that is missing the entire display
assembly and it works just fine over GPIB. Try it, it would most likely
work.
Didier KO4BB
On Thu, Oct 22, 2020, 9:46 AM Mark Spencer mark@alignedsolutions.com
wrote:
I am just curious how viable would it be to use a 531xx series counter
with Timelab if the counter display has failed ?
Thanks in advance
Mark Spencer
mark@alignedsolutions.com
604 762 4099
Hi,
As a newbie in the field, I've collected, by chance, some frequency
Now is the time to measure them, and see how they perform.
I've inherited a 5371, but something tells me that is not the right
instrument - and maybe is even faulty.
So, I'm in the need to buy a counter to use togheter with TimeLab.
Choices are many, but I'll try to avoid the "older" machines lile 5370
or 5335. The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?
53181 = single channel 400 ps counter
53131 = dual channel 400 ps counter
53132 = dual channel 100 ps counter
5370 (when working) = dual channel 20 ps counter
The gotcha is that none of them are good enough to properly measure 1
second ADEV on a good ( but still could be cheap ) OCXO. A mixer based
setup is a cheap way to get things done, even with a 5335.
The 531xx counters all share a couple of issues:
-
The power supply was made by who knows who and their quality is not
the best. Good news is that the power supply out of a (cheap) 53181 will
swap over to revive an expensive 53132.
-
They have a fan, it plugs with dust. When that happens it gets hot
the cabinet. Things (like the power supply) die as a result.
- They have a VFD display. It’s a nice one, but they do wear out.
them is problematic.
- They have a fancy setup to add digits to a frequency measurement. It
can lead you astray. It also tends to go deaf right at 10 MHz. There are
notes out there that explain the details.
- With any counter, a good external reference is the way to go for
grade measurements.
None of that is to say they are a bad counter, far from it. I have a
of them and have been using them here and at work for decades. The 532xx
counters are the latest and greatest. They are on eBay, but not cheap.
Bob
Can somebody shed some light, and maybe help even to found a baseline
and follow the instructions there.
and follow the instructions there.
Just for reference, I have an HP5334 that is missing the entire display
assembly and it works just fine over GPIB. Try it, it would most likely
work.
Didier KO4BB
On Thu, Oct 22, 2020, 9:46 AM Mark Spencer <mark@alignedsolutions.com>
wrote:
> I am just curious how viable would it be to use a 531xx series counter
> with Timelab if the counter display has failed ?
>
> Thanks in advance
>
> Mark Spencer
> mark@alignedsolutions.com
> 604 762 4099
>
> > On Oct 22, 2020, at 6:36 AM, Bob kb8tq <kb8tq@n1k.org> wrote:
> >
> > Hi
> >
> >> On Oct 22, 2020, at 7:50 AM, Giorgio Barinetti <giorgio@barinetti.it>
> wrote:
> >>
> >> Hi,
> >>
> >> As a newbie in the field, I've collected, by chance, some frequency
> standards.
> >>
> >> Now is the time to measure them, and see how they perform.
> >>
> >> I've inherited a 5371, but something tells me that is not the right
> instrument - and maybe is even faulty.
> >>
> >> So, I'm in the need to buy a counter to use togheter with TimeLab.
> >>
> >> Choices are many, but I'll try to avoid the "older" machines lile 5370
> or 5335. The 531xx series seems nice ( money apart )
> >> But again : which one between the 3 ? 53131, 53132 or 53181 ?
> >
> > 53181 = single channel 400 ps counter
> > 53131 = dual channel 400 ps counter
> > 53132 = dual channel 100 ps counter
> >
> > 5370 (when working) = dual channel 20 ps counter
> >
> > The gotcha is that none of them are good enough to properly measure 1
> > second ADEV on a good ( but still could be cheap ) OCXO. A mixer based
> > setup is a cheap way to get things done, even with a 5335.
> >
> > The 531xx counters all share a couple of issues:
> >
> > 1) The power supply was made by who knows who and their quality is not
> > the best. Good news is that the power supply out of a (cheap) 53181 will
> > swap over to revive an expensive 53132.
> >
> > 2) They have a fan, it plugs with dust. When that happens it gets hot
> inside
> > the cabinet. Things (like the power supply) die as a result.
> >
> > 3) They have a VFD display. It’s a nice one, but they do wear out.
> Replacing
> > them is problematic.
> >
> > 4) They have a fancy setup to add digits to a frequency measurement. It
> > can lead you astray. It also tends to go deaf right at 10 MHz. There are
> app
> > notes out there that explain the details.
> >
> > 5) With *any* counter, a good external reference is the way to go for
> “Time Nut”
> > grade measurements.
> >
> > None of that is to say they are a bad counter, far from it. I have a
> number
> > of them and have been using them here and at work for decades. The 532xx
> > counters are the latest and greatest. They are on eBay, but not cheap.
> >
> > Bob
> >
> >
> >
> >>
> >> Can somebody shed some light, and maybe help even to found a baseline
> for us beginners ?
> >>
> >> Regards,
> >> IZ2JGB
> >> Giorgio
> >>
> >> My NTP servers: (ntpd on FreeBSD - PPS in via Serial)
> >>
> >> HP/Symmetricom 55300A https://www.ntppool.org/scores/93.41.196.243
> >>
> >> Efratom Rb/Xc GPSDO https://www.ntppool.org/scores/95.255.136.126
> >>
> >>
> >>
> >> _______________________________________________
> >> time-nuts mailing list -- time-nuts@lists.febo.com
> >> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >> and follow the instructions there.
> >
> >
> > _______________________________________________
> > time-nuts mailing list -- time-nuts@lists.febo.com
> > To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> > and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
>
R
Rex
Thu, Nov 5, 2020 8:30 AM
Hi Magnus,
Just catching up on list messages and saw this one from you.
I have a Fluke PM 6881 counter. I don't think I've ever seen a
description of a method for reading/restoring these battery backed up
calibration constants. I looked for a way, as losing them is something
I've worried about. Not that it has happened and I did replace the
battery once.
Is doing this described in one of the manuals? Sounds like it is through
GPIB? I'd greatly appreciate any pointers to info or other details you
might provide.
I did see, in the service manual, a short description of a method for
Interpolator calibration that seems to be for making these calibrations.
Seems if the saved cal values get lost, the counter will display
"CaL.LOSt". The cal procedure is driven by an old DOS program (that I've
never found) and requires a: PM5768 Pulse gen, PM5193 LF Sig Gen, good
10 MHz, all GPIB controlled from the program. Never saw more details but
sounds messy. If there is description of GPIB commands for
reading/setting cal values, I missed them.
thanks for mentioning this and anything more you can provide
-Rex
On 10/29/2020 5:37 PM, Magnus Danielson wrote:
Hi,
I second this. You can read the calibration data out of the counter and
save. I've done some experiments with that, but nothing conclusive, but
I blame my lack of patience and not a proper setup.
Do replace the battery, it is cheap and relatively easy to do.
Would you loose this calibration, through a little GPIB commands one can
write a fake value in. This will however not produce the best resu. The
calibration routine actually runs an off beat frequency and then test
different values, and look for least RMS value, because it is the
calibration point. I've not had time to replicate all that, but I did
manage to write the fake value in and at one time get rid of the CAL
LOST warning.
Cheers,
Magnus
On 2020-10-29 14:17, Azelio Boriani wrote:
For those who have the PM6681 (aka CNT81): check the 3V memory backup
cell and replace it before the dreaded calibration lost (CAL LOST)
will appear on the LCD. Replace the coin cell with great care (with
the counter powered up), see the service manual for the procedure.
https://archive.org/details/FLUKE_PM6681_Service_Manual
On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson magnus@rubidium.se wrote:
Hi,
On 2020-10-22 19:13, Attila Kinali wrote:
Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?
Maybe try to get hold of one of the Philips (later licensed to
Fluke) PM6680 or PM6681? These are more common in Europe than
in the US, so the big US dominated websites/forums/.. don't
mention them that often. Solid devices that can be had as low
as 300€ if you are willing to wait, 500-800€ is the usual going
price. The SR620 is the workhorse that drives a lot of the
time and frequency metrology worldwide and can be had new and
used (new on http://thinksrs.com goes for 800-2000€ used).
If you go for a new one, I would consider looking at the
Pendulum CNT-90 and CNT-91. (Pendulum is the company that
took over Philips frequency counter business and the CNT-90
is the continuation of the PM668x line, also sold as PM6690
by Fluke)
Let me correct on the history and geniology there.
Philips had a instrument making side called Philips Industrier Järfälla
that did a range of measurement instruments. Later they joined forces
with Fluke. Later Philips felt that the business unit was a bad fit to
stay in Philips, so they sold it off to become a separate company which
became Pendelum. Pendelum was really the business unit with people etc
through that process, and the Fluke relation and rebranding continued.
Naturally Pendelum moved out of the Philips Industrier Järfälla office
over to Bälstabro (both locations in north of Stockholm) as it was sold
off. Pendelum also managed to rebrand their counters to Tektronix, which
mainly consisted of cosmetic changes to get the look and feel. Pendelum
was operated for many years like this, some of their production in
Pajala, where as other where done in Bälstabro. Later they reshaped the
production so that it moved to Poland where it remains. Pendulum was
sold to Spectracom and was operated as a subsidary for a while, until
they shut operation down.
The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
had the timing ASIC. The CNT-90 (100 ps) was developed to the CNT-91 (50
ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
CNT-92, but could not at that time do it with the same technical setup.
They also had the Wander Meter WM-10 which aided in testing telecom
sync. After some testing, I suggested they would broaden the product to
handle more signals and that is when they mostly firmware upgraded it to
the WM-11, and first time I tested the WM-11 the second '1' was cut out
from another '10' print and put over the '0'. Also, it was still under
development. Several years later they where still going back to my list
of feedback. They later did a revamped this into a new instrument with
even improved capabilities such as interfaces boards. This was later
sold off to Calnex which sells it as Sentinel, which is a great product
for telecom operators.
I still have contact with some of the Philips/Fluke/Pendelum staff and
owner.
Cheers,
Magnus
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi Magnus,
Just catching up on list messages and saw this one from you.
I have a Fluke PM 6881 counter. I don't think I've ever seen a
description of a method for reading/restoring these battery backed up
calibration constants. I looked for a way, as losing them is something
I've worried about. Not that it has happened and I did replace the
battery once.
Is doing this described in one of the manuals? Sounds like it is through
GPIB? I'd greatly appreciate any pointers to info or other details you
might provide.
I did see, in the service manual, a short description of a method for
Interpolator calibration that seems to be for making these calibrations.
Seems if the saved cal values get lost, the counter will display
"CaL.LOSt". The cal procedure is driven by an old DOS program (that I've
never found) and requires a: PM5768 Pulse gen, PM5193 LF Sig Gen, good
10 MHz, all GPIB controlled from the program. Never saw more details but
sounds messy. If there is description of GPIB commands for
reading/setting cal values, I missed them.
thanks for mentioning this and anything more you can provide
-Rex
On 10/29/2020 5:37 PM, Magnus Danielson wrote:
> Hi,
>
> I second this. You can read the calibration data out of the counter and
> save. I've done some experiments with that, but nothing conclusive, but
> I blame my lack of patience and not a proper setup.
>
> Do replace the battery, it is cheap and relatively easy to do.
>
> Would you loose this calibration, through a little GPIB commands one can
> write a fake value in. This will however not produce the best resu. The
> calibration routine actually runs an off beat frequency and then test
> different values, and look for least RMS value, because it is the
> calibration point. I've not had time to replicate all that, but I did
> manage to write the fake value in and at one time get rid of the CAL
> LOST warning.
>
> Cheers,
> Magnus
>
> On 2020-10-29 14:17, Azelio Boriani wrote:
>> For those who have the PM6681 (aka CNT81): check the 3V memory backup
>> cell and replace it before the dreaded calibration lost (CAL LOST)
>> will appear on the LCD. Replace the coin cell with great care (with
>> the counter powered up), see the service manual for the procedure.
>> <https://archive.org/details/FLUKE_PM6681_Service_Manual>
>>
>> On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson <magnus@rubidium.se> wrote:
>>> Hi,
>>>
>>> On 2020-10-22 19:13, Attila Kinali wrote:
>>>> On Thu, 22 Oct 2020 11:50:08 +0000
>>>> Giorgio Barinetti <giorgio@barinetti.it> wrote:
>>>>
>>>>> Choices are many, but I'll try to avoid the "older" machines lile 5370 or 5335. The 531xx series seems nice ( money apart )
>>>>> But again : which one between the 3 ? 53131, 53132 or 53181 ?
>>>> Maybe try to get hold of one of the Philips (later licensed to
>>>> Fluke) PM6680 or PM6681? These are more common in Europe than
>>>> in the US, so the big US dominated websites/forums/.. don't
>>>> mention them that often. Solid devices that can be had as low
>>>> as 300€ if you are willing to wait, 500-800€ is the usual going
>>>> price. The SR620 is the workhorse that drives a lot of the
>>>> time and frequency metrology worldwide and can be had new and
>>>> used (new on http://thinksrs.com goes for 800-2000€ used).
>>>>
>>>> If you go for a new one, I would consider looking at the
>>>> Pendulum CNT-90 and CNT-91. (Pendulum is the company that
>>>> took over Philips frequency counter business and the CNT-90
>>>> is the continuation of the PM668x line, also sold as PM6690
>>>> by Fluke)
>>> Let me correct on the history and geniology there.
>>>
>>> Philips had a instrument making side called Philips Industrier Järfälla
>>> that did a range of measurement instruments. Later they joined forces
>>> with Fluke. Later Philips felt that the business unit was a bad fit to
>>> stay in Philips, so they sold it off to become a separate company which
>>> became Pendelum. Pendelum was really the business unit with people etc
>>> through that process, and the Fluke relation and rebranding continued.
>>> Naturally Pendelum moved out of the Philips Industrier Järfälla office
>>> over to Bälstabro (both locations in north of Stockholm) as it was sold
>>> off. Pendelum also managed to rebrand their counters to Tektronix, which
>>> mainly consisted of cosmetic changes to get the look and feel. Pendelum
>>> was operated for many years like this, some of their production in
>>> Pajala, where as other where done in Bälstabro. Later they reshaped the
>>> production so that it moved to Poland where it remains. Pendulum was
>>> sold to Spectracom and was operated as a subsidary for a while, until
>>> they shut operation down.
>>>
>>> The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
>>> had the timing ASIC. The CNT-90 (100 ps) was developed to the CNT-91 (50
>>> ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
>>> CNT-92, but could not at that time do it with the same technical setup.
>>> They also had the Wander Meter WM-10 which aided in testing telecom
>>> sync. After some testing, I suggested they would broaden the product to
>>> handle more signals and that is when they mostly firmware upgraded it to
>>> the WM-11, and first time I tested the WM-11 the second '1' was cut out
>>> from another '10' print and put over the '0'. Also, it was still under
>>> development. Several years later they where still going back to my list
>>> of feedback. They later did a revamped this into a new instrument with
>>> even improved capabilities such as interfaces boards. This was later
>>> sold off to Calnex which sells it as Sentinel, which is a great product
>>> for telecom operators.
>>>
>>> I still have contact with some of the Philips/Fluke/Pendelum staff and
>>> owner.
>>>
>>> Cheers,
>>> Magnus
>>>
>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
>
>
MD
Magnus Danielson
Sat, Nov 7, 2020 2:08 AM
Hi Rex,
I need to dig in the archive to refresh my memory. I don't recall
precisely, but I think I recalled that the manual indirectly describes
the calibration data string.
I have learned a few things from Pendulum, but I did not have the right
tools at hand to set things up.
There was a more recent setup that could use more modern generators, but
the trick was still the same. You lock the generator and counter to the
same frequency, then you set the generator to a small offset frequency
from 10 MHz, which is 9.999 MHz as I recall it. This slowly sweeps
through all the phase-relationships between the reference oscillator and
the counter input, thus sweeps the interpolator phase. It then chooses
the calibration constant giving the lowest RMS error, as this is the
best compensation for the hardware min-point. All this is free from
memory. Then that value with calibration date is written into memory. If
I recall correctly 2.21 ns is a typical value.
I have PM6681 in need of calibration, and as I recall it I was able to
program it enough for the calibration error warning did not show up.
This not to say it was actually calibrated.
At some point I will return to that project. The generator I used did
not support that offset frequency, but I have others that do. Also, my
crapiola GPIB programming needs attention. My intention is to share the
fruits of this project when it comes to that. The lab has been in
shambles for too long, but shaping up slowly and nicely.
Cheers,
Magnus
On 2020-11-05 09:30, Rex wrote:
Hi Magnus,
Just catching up on list messages and saw this one from you.
I have a Fluke PM 6881 counter. I don't think I've ever seen a
description of a method for reading/restoring these battery backed up
calibration constants. I looked for a way, as losing them is something
I've worried about. Not that it has happened and I did replace the
battery once.
Is doing this described in one of the manuals? Sounds like it is
through GPIB? I'd greatly appreciate any pointers to info or other
details you might provide.
I did see, in the service manual, a short description of a method for
Interpolator calibration that seems to be for making these
calibrations. Seems if the saved cal values get lost, the counter will
display "CaL.LOSt". The cal procedure is driven by an old DOS program
(that I've never found) and requires a: PM5768 Pulse gen, PM5193 LF
Sig Gen, good 10 MHz, all GPIB controlled from the program. Never saw
more details but sounds messy. If there is description of GPIB
commands for reading/setting cal values, I missed them.
thanks for mentioning this and anything more you can provide
-Rex
On 10/29/2020 5:37 PM, Magnus Danielson wrote:
Hi,
I second this. You can read the calibration data out of the counter and
save. I've done some experiments with that, but nothing conclusive, but
I blame my lack of patience and not a proper setup.
Do replace the battery, it is cheap and relatively easy to do.
Would you loose this calibration, through a little GPIB commands one can
write a fake value in. This will however not produce the best resu. The
calibration routine actually runs an off beat frequency and then test
different values, and look for least RMS value, because it is the
calibration point. I've not had time to replicate all that, but I did
manage to write the fake value in and at one time get rid of the CAL
LOST warning.
Cheers,
Magnus
On 2020-10-29 14:17, Azelio Boriani wrote:
For those who have the PM6681 (aka CNT81): check the 3V memory backup
cell and replace it before the dreaded calibration lost (CAL LOST)
will appear on the LCD. Replace the coin cell with great care (with
the counter powered up), see the service manual for the procedure.
https://archive.org/details/FLUKE_PM6681_Service_Manual
On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson
magnus@rubidium.se wrote:
Hi,
On 2020-10-22 19:13, Attila Kinali wrote:
Choices are many, but I'll try to avoid the "older" machines lile
5370 or 5335. The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?
Maybe try to get hold of one of the Philips (later licensed to
Fluke) PM6680 or PM6681? These are more common in Europe than
in the US, so the big US dominated websites/forums/.. don't
mention them that often. Solid devices that can be had as low
as 300€ if you are willing to wait, 500-800€ is the usual going
price. The SR620 is the workhorse that drives a lot of the
time and frequency metrology worldwide and can be had new and
used (new on http://thinksrs.com goes for 800-2000€ used).
If you go for a new one, I would consider looking at the
Pendulum CNT-90 and CNT-91. (Pendulum is the company that
took over Philips frequency counter business and the CNT-90
is the continuation of the PM668x line, also sold as PM6690
by Fluke)
Let me correct on the history and geniology there.
Philips had a instrument making side called Philips Industrier
Järfälla
that did a range of measurement instruments. Later they joined forces
with Fluke. Later Philips felt that the business unit was a bad fit to
stay in Philips, so they sold it off to become a separate company
which
became Pendelum. Pendelum was really the business unit with people etc
through that process, and the Fluke relation and rebranding continued.
Naturally Pendelum moved out of the Philips Industrier Järfälla office
over to Bälstabro (both locations in north of Stockholm) as it was
sold
off. Pendelum also managed to rebrand their counters to Tektronix,
which
mainly consisted of cosmetic changes to get the look and feel.
Pendelum
was operated for many years like this, some of their production in
Pajala, where as other where done in Bälstabro. Later they reshaped
the
production so that it moved to Poland where it remains. Pendulum was
sold to Spectracom and was operated as a subsidary for a while, until
they shut operation down.
The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
had the timing ASIC. The CNT-90 (100 ps) was developed to the
CNT-91 (50
ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
CNT-92, but could not at that time do it with the same technical
setup.
They also had the Wander Meter WM-10 which aided in testing telecom
sync. After some testing, I suggested they would broaden the
product to
handle more signals and that is when they mostly firmware upgraded
it to
the WM-11, and first time I tested the WM-11 the second '1' was cut
out
from another '10' print and put over the '0'. Also, it was still under
development. Several years later they where still going back to my
list
of feedback. They later did a revamped this into a new instrument with
even improved capabilities such as interfaces boards. This was later
sold off to Calnex which sells it as Sentinel, which is a great
product
for telecom operators.
I still have contact with some of the Philips/Fluke/Pendelum staff and
owner.
Cheers,
Magnus
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Hi Rex,
I need to dig in the archive to refresh my memory. I don't recall
precisely, but I think I recalled that the manual indirectly describes
the calibration data string.
I have learned a few things from Pendulum, but I did not have the right
tools at hand to set things up.
There was a more recent setup that could use more modern generators, but
the trick was still the same. You lock the generator and counter to the
same frequency, then you set the generator to a small offset frequency
from 10 MHz, which is 9.999 MHz as I recall it. This slowly sweeps
through all the phase-relationships between the reference oscillator and
the counter input, thus sweeps the interpolator phase. It then chooses
the calibration constant giving the lowest RMS error, as this is the
best compensation for the hardware min-point. All this is free from
memory. Then that value with calibration date is written into memory. If
I recall correctly 2.21 ns is a typical value.
I have PM6681 in need of calibration, and as I recall it I was able to
program it enough for the calibration error warning did not show up.
This not to say it was actually calibrated.
At some point I will return to that project. The generator I used did
not support that offset frequency, but I have others that do. Also, my
crapiola GPIB programming needs attention. My intention is to share the
fruits of this project when it comes to that. The lab has been in
shambles for too long, but shaping up slowly and nicely.
Cheers,
Magnus
On 2020-11-05 09:30, Rex wrote:
> Hi Magnus,
>
> Just catching up on list messages and saw this one from you.
>
> I have a Fluke PM 6881 counter. I don't think I've ever seen a
> description of a method for reading/restoring these battery backed up
> calibration constants. I looked for a way, as losing them is something
> I've worried about. Not that it has happened and I did replace the
> battery once.
>
> Is doing this described in one of the manuals? Sounds like it is
> through GPIB? I'd greatly appreciate any pointers to info or other
> details you might provide.
>
> I did see, in the service manual, a short description of a method for
> Interpolator calibration that seems to be for making these
> calibrations. Seems if the saved cal values get lost, the counter will
> display "CaL.LOSt". The cal procedure is driven by an old DOS program
> (that I've never found) and requires a: PM5768 Pulse gen, PM5193 LF
> Sig Gen, good 10 MHz, all GPIB controlled from the program. Never saw
> more details but sounds messy. If there is description of GPIB
> commands for reading/setting cal values, I missed them.
>
> thanks for mentioning this and anything more you can provide
> -Rex
>
> On 10/29/2020 5:37 PM, Magnus Danielson wrote:
>> Hi,
>>
>> I second this. You can read the calibration data out of the counter and
>> save. I've done some experiments with that, but nothing conclusive, but
>> I blame my lack of patience and not a proper setup.
>>
>> Do replace the battery, it is cheap and relatively easy to do.
>>
>> Would you loose this calibration, through a little GPIB commands one can
>> write a fake value in. This will however not produce the best resu. The
>> calibration routine actually runs an off beat frequency and then test
>> different values, and look for least RMS value, because it is the
>> calibration point. I've not had time to replicate all that, but I did
>> manage to write the fake value in and at one time get rid of the CAL
>> LOST warning.
>>
>> Cheers,
>> Magnus
>>
>> On 2020-10-29 14:17, Azelio Boriani wrote:
>>> For those who have the PM6681 (aka CNT81): check the 3V memory backup
>>> cell and replace it before the dreaded calibration lost (CAL LOST)
>>> will appear on the LCD. Replace the coin cell with great care (with
>>> the counter powered up), see the service manual for the procedure.
>>> <https://archive.org/details/FLUKE_PM6681_Service_Manual>
>>>
>>> On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson
>>> <magnus@rubidium.se> wrote:
>>>> Hi,
>>>>
>>>> On 2020-10-22 19:13, Attila Kinali wrote:
>>>>> On Thu, 22 Oct 2020 11:50:08 +0000
>>>>> Giorgio Barinetti <giorgio@barinetti.it> wrote:
>>>>>
>>>>>> Choices are many, but I'll try to avoid the "older" machines lile
>>>>>> 5370 or 5335. The 531xx series seems nice ( money apart )
>>>>>> But again : which one between the 3 ? 53131, 53132 or 53181 ?
>>>>> Maybe try to get hold of one of the Philips (later licensed to
>>>>> Fluke) PM6680 or PM6681? These are more common in Europe than
>>>>> in the US, so the big US dominated websites/forums/.. don't
>>>>> mention them that often. Solid devices that can be had as low
>>>>> as 300€ if you are willing to wait, 500-800€ is the usual going
>>>>> price. The SR620 is the workhorse that drives a lot of the
>>>>> time and frequency metrology worldwide and can be had new and
>>>>> used (new on http://thinksrs.com goes for 800-2000€ used).
>>>>>
>>>>> If you go for a new one, I would consider looking at the
>>>>> Pendulum CNT-90 and CNT-91. (Pendulum is the company that
>>>>> took over Philips frequency counter business and the CNT-90
>>>>> is the continuation of the PM668x line, also sold as PM6690
>>>>> by Fluke)
>>>> Let me correct on the history and geniology there.
>>>>
>>>> Philips had a instrument making side called Philips Industrier
>>>> Järfälla
>>>> that did a range of measurement instruments. Later they joined forces
>>>> with Fluke. Later Philips felt that the business unit was a bad fit to
>>>> stay in Philips, so they sold it off to become a separate company
>>>> which
>>>> became Pendelum. Pendelum was really the business unit with people etc
>>>> through that process, and the Fluke relation and rebranding continued.
>>>> Naturally Pendelum moved out of the Philips Industrier Järfälla office
>>>> over to Bälstabro (both locations in north of Stockholm) as it was
>>>> sold
>>>> off. Pendelum also managed to rebrand their counters to Tektronix,
>>>> which
>>>> mainly consisted of cosmetic changes to get the look and feel.
>>>> Pendelum
>>>> was operated for many years like this, some of their production in
>>>> Pajala, where as other where done in Bälstabro. Later they reshaped
>>>> the
>>>> production so that it moved to Poland where it remains. Pendulum was
>>>> sold to Spectracom and was operated as a subsidary for a while, until
>>>> they shut operation down.
>>>>
>>>> The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
>>>> had the timing ASIC. The CNT-90 (100 ps) was developed to the
>>>> CNT-91 (50
>>>> ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
>>>> CNT-92, but could not at that time do it with the same technical
>>>> setup.
>>>> They also had the Wander Meter WM-10 which aided in testing telecom
>>>> sync. After some testing, I suggested they would broaden the
>>>> product to
>>>> handle more signals and that is when they mostly firmware upgraded
>>>> it to
>>>> the WM-11, and first time I tested the WM-11 the second '1' was cut
>>>> out
>>>> from another '10' print and put over the '0'. Also, it was still under
>>>> development. Several years later they where still going back to my
>>>> list
>>>> of feedback. They later did a revamped this into a new instrument with
>>>> even improved capabilities such as interfaces boards. This was later
>>>> sold off to Calnex which sells it as Sentinel, which is a great
>>>> product
>>>> for telecom operators.
>>>>
>>>> I still have contact with some of the Philips/Fluke/Pendelum staff and
>>>> owner.
>>>>
>>>> Cheers,
>>>> Magnus
>>>>
>>>>
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>>> To unsubscribe, go to
>>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>>>> and follow the instructions there.
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@lists.febo.com
>>> To unsubscribe, go to
>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>>> and follow the instructions there.
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe, go to
>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>> and follow the instructions there.
>>
>>
>
AB
Azelio Boriani
Sun, Nov 8, 2020 10:01 PM
Old story about the PM6681 (18 Nov 2015, thread: "PM6681 and Timelab")
where a sort of calibration procedure is described: the PUD command is
NOT a calibration command.
PM6681 programming manual, page 9-127: PUD Protected User Data...This
is a data area where the user may write ANY data up to 64
characters...
If the user can write any data, how can it be a calibration command or
calibration data area?
Better watch out those 3V coin cells, we will never get the real
calibration commands/procedure. I have tried with the disassembled
firmware, no way. The visible strings of GPIB commands are all
described in the programming manual, so nothing useful.
On Sat, Nov 7, 2020 at 3:31 AM Magnus Danielson magnus@rubidium.se wrote:
Hi Rex,
I need to dig in the archive to refresh my memory. I don't recall
precisely, but I think I recalled that the manual indirectly describes
the calibration data string.
I have learned a few things from Pendulum, but I did not have the right
tools at hand to set things up.
There was a more recent setup that could use more modern generators, but
the trick was still the same. You lock the generator and counter to the
same frequency, then you set the generator to a small offset frequency
from 10 MHz, which is 9.999 MHz as I recall it. This slowly sweeps
through all the phase-relationships between the reference oscillator and
the counter input, thus sweeps the interpolator phase. It then chooses
the calibration constant giving the lowest RMS error, as this is the
best compensation for the hardware min-point. All this is free from
memory. Then that value with calibration date is written into memory. If
I recall correctly 2.21 ns is a typical value.
I have PM6681 in need of calibration, and as I recall it I was able to
program it enough for the calibration error warning did not show up.
This not to say it was actually calibrated.
At some point I will return to that project. The generator I used did
not support that offset frequency, but I have others that do. Also, my
crapiola GPIB programming needs attention. My intention is to share the
fruits of this project when it comes to that. The lab has been in
shambles for too long, but shaping up slowly and nicely.
Cheers,
Magnus
On 2020-11-05 09:30, Rex wrote:
Hi Magnus,
Just catching up on list messages and saw this one from you.
I have a Fluke PM 6881 counter. I don't think I've ever seen a
description of a method for reading/restoring these battery backed up
calibration constants. I looked for a way, as losing them is something
I've worried about. Not that it has happened and I did replace the
battery once.
Is doing this described in one of the manuals? Sounds like it is
through GPIB? I'd greatly appreciate any pointers to info or other
details you might provide.
I did see, in the service manual, a short description of a method for
Interpolator calibration that seems to be for making these
calibrations. Seems if the saved cal values get lost, the counter will
display "CaL.LOSt". The cal procedure is driven by an old DOS program
(that I've never found) and requires a: PM5768 Pulse gen, PM5193 LF
Sig Gen, good 10 MHz, all GPIB controlled from the program. Never saw
more details but sounds messy. If there is description of GPIB
commands for reading/setting cal values, I missed them.
thanks for mentioning this and anything more you can provide
-Rex
On 10/29/2020 5:37 PM, Magnus Danielson wrote:
Hi,
I second this. You can read the calibration data out of the counter and
save. I've done some experiments with that, but nothing conclusive, but
I blame my lack of patience and not a proper setup.
Do replace the battery, it is cheap and relatively easy to do.
Would you loose this calibration, through a little GPIB commands one can
write a fake value in. This will however not produce the best resu. The
calibration routine actually runs an off beat frequency and then test
different values, and look for least RMS value, because it is the
calibration point. I've not had time to replicate all that, but I did
manage to write the fake value in and at one time get rid of the CAL
LOST warning.
Cheers,
Magnus
On 2020-10-29 14:17, Azelio Boriani wrote:
For those who have the PM6681 (aka CNT81): check the 3V memory backup
cell and replace it before the dreaded calibration lost (CAL LOST)
will appear on the LCD. Replace the coin cell with great care (with
the counter powered up), see the service manual for the procedure.
https://archive.org/details/FLUKE_PM6681_Service_Manual
On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson
magnus@rubidium.se wrote:
Hi,
On 2020-10-22 19:13, Attila Kinali wrote:
Choices are many, but I'll try to avoid the "older" machines lile
5370 or 5335. The 531xx series seems nice ( money apart )
But again : which one between the 3 ? 53131, 53132 or 53181 ?
Maybe try to get hold of one of the Philips (later licensed to
Fluke) PM6680 or PM6681? These are more common in Europe than
in the US, so the big US dominated websites/forums/.. don't
mention them that often. Solid devices that can be had as low
as 300€ if you are willing to wait, 500-800€ is the usual going
price. The SR620 is the workhorse that drives a lot of the
time and frequency metrology worldwide and can be had new and
used (new on http://thinksrs.com goes for 800-2000€ used).
If you go for a new one, I would consider looking at the
Pendulum CNT-90 and CNT-91. (Pendulum is the company that
took over Philips frequency counter business and the CNT-90
is the continuation of the PM668x line, also sold as PM6690
by Fluke)
Let me correct on the history and geniology there.
Philips had a instrument making side called Philips Industrier
Järfälla
that did a range of measurement instruments. Later they joined forces
with Fluke. Later Philips felt that the business unit was a bad fit to
stay in Philips, so they sold it off to become a separate company
which
became Pendelum. Pendelum was really the business unit with people etc
through that process, and the Fluke relation and rebranding continued.
Naturally Pendelum moved out of the Philips Industrier Järfälla office
over to Bälstabro (both locations in north of Stockholm) as it was
sold
off. Pendelum also managed to rebrand their counters to Tektronix,
which
mainly consisted of cosmetic changes to get the look and feel.
Pendelum
was operated for many years like this, some of their production in
Pajala, where as other where done in Bälstabro. Later they reshaped
the
production so that it moved to Poland where it remains. Pendulum was
sold to Spectracom and was operated as a subsidary for a while, until
they shut operation down.
The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
had the timing ASIC. The CNT-90 (100 ps) was developed to the
CNT-91 (50
ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
CNT-92, but could not at that time do it with the same technical
setup.
They also had the Wander Meter WM-10 which aided in testing telecom
sync. After some testing, I suggested they would broaden the
product to
handle more signals and that is when they mostly firmware upgraded
it to
the WM-11, and first time I tested the WM-11 the second '1' was cut
out
from another '10' print and put over the '0'. Also, it was still under
development. Several years later they where still going back to my
list
of feedback. They later did a revamped this into a new instrument with
even improved capabilities such as interfaces boards. This was later
sold off to Calnex which sells it as Sentinel, which is a great
product
for telecom operators.
I still have contact with some of the Philips/Fluke/Pendelum staff and
owner.
Cheers,
Magnus
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.
Old story about the PM6681 (18 Nov 2015, thread: "PM6681 and Timelab")
where a sort of calibration procedure is described: the PUD command is
NOT a calibration command.
PM6681 programming manual, page 9-127: PUD Protected User Data...This
is a data area where the user may write ANY data up to 64
characters...
If the user can write any data, how can it be a calibration command or
calibration data area?
Better watch out those 3V coin cells, we will never get the real
calibration commands/procedure. I have tried with the disassembled
firmware, no way. The visible strings of GPIB commands are all
described in the programming manual, so nothing useful.
On Sat, Nov 7, 2020 at 3:31 AM Magnus Danielson <magnus@rubidium.se> wrote:
>
> Hi Rex,
>
> I need to dig in the archive to refresh my memory. I don't recall
> precisely, but I think I recalled that the manual indirectly describes
> the calibration data string.
>
> I have learned a few things from Pendulum, but I did not have the right
> tools at hand to set things up.
>
> There was a more recent setup that could use more modern generators, but
> the trick was still the same. You lock the generator and counter to the
> same frequency, then you set the generator to a small offset frequency
> from 10 MHz, which is 9.999 MHz as I recall it. This slowly sweeps
> through all the phase-relationships between the reference oscillator and
> the counter input, thus sweeps the interpolator phase. It then chooses
> the calibration constant giving the lowest RMS error, as this is the
> best compensation for the hardware min-point. All this is free from
> memory. Then that value with calibration date is written into memory. If
> I recall correctly 2.21 ns is a typical value.
>
> I have PM6681 in need of calibration, and as I recall it I was able to
> program it enough for the calibration error warning did not show up.
> This not to say it was actually calibrated.
>
> At some point I will return to that project. The generator I used did
> not support that offset frequency, but I have others that do. Also, my
> crapiola GPIB programming needs attention. My intention is to share the
> fruits of this project when it comes to that. The lab has been in
> shambles for too long, but shaping up slowly and nicely.
>
> Cheers,
> Magnus
>
> On 2020-11-05 09:30, Rex wrote:
> > Hi Magnus,
> >
> > Just catching up on list messages and saw this one from you.
> >
> > I have a Fluke PM 6881 counter. I don't think I've ever seen a
> > description of a method for reading/restoring these battery backed up
> > calibration constants. I looked for a way, as losing them is something
> > I've worried about. Not that it has happened and I did replace the
> > battery once.
> >
> > Is doing this described in one of the manuals? Sounds like it is
> > through GPIB? I'd greatly appreciate any pointers to info or other
> > details you might provide.
> >
> > I did see, in the service manual, a short description of a method for
> > Interpolator calibration that seems to be for making these
> > calibrations. Seems if the saved cal values get lost, the counter will
> > display "CaL.LOSt". The cal procedure is driven by an old DOS program
> > (that I've never found) and requires a: PM5768 Pulse gen, PM5193 LF
> > Sig Gen, good 10 MHz, all GPIB controlled from the program. Never saw
> > more details but sounds messy. If there is description of GPIB
> > commands for reading/setting cal values, I missed them.
> >
> > thanks for mentioning this and anything more you can provide
> > -Rex
> >
> > On 10/29/2020 5:37 PM, Magnus Danielson wrote:
> >> Hi,
> >>
> >> I second this. You can read the calibration data out of the counter and
> >> save. I've done some experiments with that, but nothing conclusive, but
> >> I blame my lack of patience and not a proper setup.
> >>
> >> Do replace the battery, it is cheap and relatively easy to do.
> >>
> >> Would you loose this calibration, through a little GPIB commands one can
> >> write a fake value in. This will however not produce the best resu. The
> >> calibration routine actually runs an off beat frequency and then test
> >> different values, and look for least RMS value, because it is the
> >> calibration point. I've not had time to replicate all that, but I did
> >> manage to write the fake value in and at one time get rid of the CAL
> >> LOST warning.
> >>
> >> Cheers,
> >> Magnus
> >>
> >> On 2020-10-29 14:17, Azelio Boriani wrote:
> >>> For those who have the PM6681 (aka CNT81): check the 3V memory backup
> >>> cell and replace it before the dreaded calibration lost (CAL LOST)
> >>> will appear on the LCD. Replace the coin cell with great care (with
> >>> the counter powered up), see the service manual for the procedure.
> >>> <https://archive.org/details/FLUKE_PM6681_Service_Manual>
> >>>
> >>> On Fri, Oct 23, 2020 at 6:17 PM Magnus Danielson
> >>> <magnus@rubidium.se> wrote:
> >>>> Hi,
> >>>>
> >>>> On 2020-10-22 19:13, Attila Kinali wrote:
> >>>>> On Thu, 22 Oct 2020 11:50:08 +0000
> >>>>> Giorgio Barinetti <giorgio@barinetti.it> wrote:
> >>>>>
> >>>>>> Choices are many, but I'll try to avoid the "older" machines lile
> >>>>>> 5370 or 5335. The 531xx series seems nice ( money apart )
> >>>>>> But again : which one between the 3 ? 53131, 53132 or 53181 ?
> >>>>> Maybe try to get hold of one of the Philips (later licensed to
> >>>>> Fluke) PM6680 or PM6681? These are more common in Europe than
> >>>>> in the US, so the big US dominated websites/forums/.. don't
> >>>>> mention them that often. Solid devices that can be had as low
> >>>>> as 300€ if you are willing to wait, 500-800€ is the usual going
> >>>>> price. The SR620 is the workhorse that drives a lot of the
> >>>>> time and frequency metrology worldwide and can be had new and
> >>>>> used (new on http://thinksrs.com goes for 800-2000€ used).
> >>>>>
> >>>>> If you go for a new one, I would consider looking at the
> >>>>> Pendulum CNT-90 and CNT-91. (Pendulum is the company that
> >>>>> took over Philips frequency counter business and the CNT-90
> >>>>> is the continuation of the PM668x line, also sold as PM6690
> >>>>> by Fluke)
> >>>> Let me correct on the history and geniology there.
> >>>>
> >>>> Philips had a instrument making side called Philips Industrier
> >>>> Järfälla
> >>>> that did a range of measurement instruments. Later they joined forces
> >>>> with Fluke. Later Philips felt that the business unit was a bad fit to
> >>>> stay in Philips, so they sold it off to become a separate company
> >>>> which
> >>>> became Pendelum. Pendelum was really the business unit with people etc
> >>>> through that process, and the Fluke relation and rebranding continued.
> >>>> Naturally Pendelum moved out of the Philips Industrier Järfälla office
> >>>> over to Bälstabro (both locations in north of Stockholm) as it was
> >>>> sold
> >>>> off. Pendelum also managed to rebrand their counters to Tektronix,
> >>>> which
> >>>> mainly consisted of cosmetic changes to get the look and feel.
> >>>> Pendelum
> >>>> was operated for many years like this, some of their production in
> >>>> Pajala, where as other where done in Bälstabro. Later they reshaped
> >>>> the
> >>>> production so that it moved to Poland where it remains. Pendulum was
> >>>> sold to Spectracom and was operated as a subsidary for a while, until
> >>>> they shut operation down.
> >>>>
> >>>> The CNT-80/81 (PM 6680 and PM6681) production went on as long as they
> >>>> had the timing ASIC. The CNT-90 (100 ps) was developed to the
> >>>> CNT-91 (50
> >>>> ps), where the later replaced the CNT-81 (50 ps). They aimed to do the
> >>>> CNT-92, but could not at that time do it with the same technical
> >>>> setup.
> >>>> They also had the Wander Meter WM-10 which aided in testing telecom
> >>>> sync. After some testing, I suggested they would broaden the
> >>>> product to
> >>>> handle more signals and that is when they mostly firmware upgraded
> >>>> it to
> >>>> the WM-11, and first time I tested the WM-11 the second '1' was cut
> >>>> out
> >>>> from another '10' print and put over the '0'. Also, it was still under
> >>>> development. Several years later they where still going back to my
> >>>> list
> >>>> of feedback. They later did a revamped this into a new instrument with
> >>>> even improved capabilities such as interfaces boards. This was later
> >>>> sold off to Calnex which sells it as Sentinel, which is a great
> >>>> product
> >>>> for telecom operators.
> >>>>
> >>>> I still have contact with some of the Philips/Fluke/Pendelum staff and
> >>>> owner.
> >>>>
> >>>> Cheers,
> >>>> Magnus
> >>>>
> >>>>
> >>>>
> >>>> _______________________________________________
> >>>> time-nuts mailing list -- time-nuts@lists.febo.com
> >>>> To unsubscribe, go to
> >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >>>> and follow the instructions there.
> >>> _______________________________________________
> >>> time-nuts mailing list -- time-nuts@lists.febo.com
> >>> To unsubscribe, go to
> >>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >>> and follow the instructions there.
> >>
> >> _______________________________________________
> >> time-nuts mailing list -- time-nuts@lists.febo.com
> >> To unsubscribe, go to
> >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >> and follow the instructions there.
> >>
> >>
> >
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.