I finally had a chance to drop my HP3458A into a friendly Cal-lab and since
the last sticker on it says it is due for calibration in "2-1996" we are
talking seriously overdue calibration.
The result:
Internal 40K resistor: -11.66ppm
Internal 7V Voltage ref: -28.96ppm
That's not too shabby over 15 years...
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
In message 2783.1315238745@critter.freebsd.dk, Poul-Henning Kamp writes:
I finally had a chance to drop my HP3458A into a friendly Cal-lab and since
the last sticker on it says it is due for calibration in "2-1996" we are
talking seriously overdue calibration.
The result:
Internal 40K resistor: -11.66ppm
Internal 7V Voltage ref: -28.96ppm
That's not too shabby over 15 years...
Update:
Greg Burnett and I have been chatting offline, and it transpires,
based on the serial number of 2823A00556, that it is more like 22
years.
Appearantly this instrument was never calibrated after it left
the factory in late 1988 or early 1989.
The stickers say "Next test" (in german), they give no hint that
a calibration has happened, and the CALNUM? was 5 when I received
the meter, consistent with the factory calibration.
Further more, we probably didn't let the meter acclimatize long
enough yesterday, so in particular the 7V cal may be a couple of
PPM off (didn't want to overstay my welcome).
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
The result:
Internal 40K resistor: -11.66ppm
Internal 7V Voltage ref: -28.96ppm
That's not too shabby over 15 years...
Update:
Greg Burnett and I have been chatting offline, and it transpires,
based on the serial number of 2823A00556, that it is more like 22
years.
Appearantly this instrument was never calibrated after it left
the factory in late 1988 or early 1989.
Hello Paul,
If the last cal was really before 1.1.1990 remember that the
official volt has been redefined at 1.1.1990 by +8.06 ppm.
So all readings which were made before have to be corrected by -8.06ppm.
With best regards
Andreas
In message 33473809FD874529A25FBD3D6B6EB55E@laptop, "Andreas Jahn" writes:
If the last cal was really before 1.1.1990 remember that the
official volt has been redefined at 1.1.1990 by +8.06 ppm.
So all readings which were made before have to be corrected by -8.06ppm.
Now things get interesting: the 8.06 was for "most other countries",
but for US it was 9.26 PPM.
Does anybody know if HP calibrated 3458A's going to europe differently
from those going to USA ?
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.