Re: [volt-nuts] Solartron 7081 Test Leads

BE
Bill Ezell
Thu, Aug 4, 2011 10:08 PM

Joe, If you don't have a cal key, the switch form-factor is standard. I
just replaced the switch on my 7081. If I remember correctly, the part
number isn't the same, but the mfgr is. I ordered from Digikey.

Who did you order the Fischer connectors from? I tracked down part
numbers a few years ago, but couldn't find anyone to actually sell them
to me. I found one at a flea market, and I did snag two original cables.

Regarding the cables/probes, there were several models. The most common
have two banana plugs. The connection thru the cable is true 4-wire,
terminating at the plug. There was also a version with 5 plugs (4-wire
plus shield), and a version with a pair of Kelvin clips (the real ones,
one of each pair to each side of the clip).

You don't need a Kelvin connection for anything other than resistance
measurement. The meter does quite well with the 2-plug version for that
anyway.

I used 5-wire + shield PTFE / pure copper (picked up on EBay, lots of
mil-spec cable shows up there, mostly from aerospace leftovers)
terminated with Pomona low-emf plugs.

BTW, these are really great meters (except for the read time at 8.5
digits!). Mine keeps cal within typically 1 or 2 ppm for over a year at
a time. Yes, seriously. I compare it periodically to my
(Fluke-calibrated) pair of Datron 4910 references, and almost never have
to recal. (Another wonderful piece of equipment, the 4910).
Nice thing about all these old pieces of equipment, they're usually well
down the aging curve. :)

One thing, they really take 24 hours to fully stabilize, so let it run a
few days if you're going to calibrate it.
Have you checked out the voltage ref it uses? Amazing.It's just a
non-ovenized Zener, albeit one that was designed to be a ref (the
still-available 1N829A). The Zener itself is good to about 5ppm/C.
Solartron then calibrated each one to find the inflection point on its
voltage/temp curve. The meter has a DAC that sets the current to be at
that point. Then, they add a non-linear temperature-dependent
compensation voltage to eliminate even more tempco. If you look at the
Chinese meter review referred to by other posters, you can see how it
compares to the HP. Pretty impressive.

One more thing, read your cal constants! The critical one is the Zener
bias setting. If you lose the EEPROM, that one would be very difficult
to determine again. The rest can be recovered by a recal.

On 08/04/2011 01:13 PM, volt-nuts-request@febo.com wrote:

Solartron 7081 Test Leads...

--
Bill Ezell

They said 'Windows or better'
so I used Linux.

Joe, If you don't have a cal key, the switch form-factor is standard. I just replaced the switch on my 7081. If I remember correctly, the part number isn't the same, but the mfgr is. I ordered from Digikey. Who did you order the Fischer connectors from? I tracked down part numbers a few years ago, but couldn't find anyone to actually sell them to me. I found one at a flea market, and I did snag two original cables. Regarding the cables/probes, there were several models. The most common have two banana plugs. The connection thru the cable is true 4-wire, terminating at the plug. There was also a version with 5 plugs (4-wire plus shield), and a version with a pair of Kelvin clips (the real ones, one of each pair to each side of the clip). You don't need a Kelvin connection for anything other than resistance measurement. The meter does quite well with the 2-plug version for that anyway. I used 5-wire + shield PTFE / pure copper (picked up on EBay, lots of mil-spec cable shows up there, mostly from aerospace leftovers) terminated with Pomona low-emf plugs. BTW, these are really great meters (except for the read time at 8.5 digits!). Mine keeps cal within typically 1 or 2 ppm for over a year at a time. Yes, seriously. I compare it periodically to my (Fluke-calibrated) pair of Datron 4910 references, and almost never have to recal. (Another wonderful piece of equipment, the 4910). Nice thing about all these old pieces of equipment, they're usually well down the aging curve. :) One thing, they really take 24 hours to fully stabilize, so let it run a few days if you're going to calibrate it. Have you checked out the voltage ref it uses? Amazing.It's just a non-ovenized Zener, albeit one that was designed to be a ref (the still-available 1N829A). The Zener itself is good to about 5ppm/C. Solartron then calibrated each one to find the inflection point on its voltage/temp curve. The meter has a DAC that sets the current to be at that point. Then, they add a non-linear temperature-dependent compensation voltage to eliminate even more tempco. If you look at the Chinese meter review referred to by other posters, you can see how it compares to the HP. Pretty impressive. One more thing, read your cal constants! The critical one is the Zener bias setting. If you lose the EEPROM, that one would be very difficult to determine again. The rest can be recovered by a recal. On 08/04/2011 01:13 PM, volt-nuts-request@febo.com wrote: > Solartron 7081 Test Leads... -- Bill Ezell ---------- They said 'Windows or better' so I used Linux.
JL
J. L. Trantham
Thu, Aug 4, 2011 10:43 PM

Bill,

Thanks for all the info.

I ordered the connectors from Kensington Electronics using the part numbers
posed on an earlier email.  Very friendly folks but they want a 'company'
name.  I used my practice.

http://www.keiconn.com/

They have a $50 minimum order.  One connector and cable clamp meets that
requirement.

I would appreciate the part number of the switch you ordered.  Also, if you
could send me a picture of the connector end and the 'probe' end of the
cables you found, I would appreciate it.

What is the EEPROM part number?  I have a couple of programmers that likely
will take care of that but I just need to know the EEPROM part number.
Also, are there other EPROM's or EEPROM's that need to be archived?

Also, are there facilities that still 'calibrate' the 7081?

I have been wondering about whether to 'twist' the wires from the connector
to the alligator clips, and, if so, which wires to 'twist' and which wires
to 'shield'.  It might make a difference with 'noise' at least as suggested
by one of the service notes I read on the 3458A.

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Bill Ezell
Sent: Thursday, August 04, 2011 5:09 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Solartron 7081 Test Leads

Joe, If you don't have a cal key, the switch form-factor is standard. I
just replaced the switch on my 7081. If I remember correctly, the part
number isn't the same, but the mfgr is. I ordered from Digikey.

Who did you order the Fischer connectors from? I tracked down part
numbers a few years ago, but couldn't find anyone to actually sell them
to me. I found one at a flea market, and I did snag two original cables.

Regarding the cables/probes, there were several models. The most common
have two banana plugs. The connection thru the cable is true 4-wire,
terminating at the plug. There was also a version with 5 plugs (4-wire
plus shield), and a version with a pair of Kelvin clips (the real ones,
one of each pair to each side of the clip).

You don't need a Kelvin connection for anything other than resistance
measurement. The meter does quite well with the 2-plug version for that
anyway.

I used 5-wire + shield PTFE / pure copper (picked up on EBay, lots of
mil-spec cable shows up there, mostly from aerospace leftovers)
terminated with Pomona low-emf plugs.

BTW, these are really great meters (except for the read time at 8.5
digits!). Mine keeps cal within typically 1 or 2 ppm for over a year at
a time. Yes, seriously. I compare it periodically to my
(Fluke-calibrated) pair of Datron 4910 references, and almost never have
to recal. (Another wonderful piece of equipment, the 4910).
Nice thing about all these old pieces of equipment, they're usually well
down the aging curve. :)

One thing, they really take 24 hours to fully stabilize, so let it run a
few days if you're going to calibrate it.
Have you checked out the voltage ref it uses? Amazing.It's just a
non-ovenized Zener, albeit one that was designed to be a ref (the
still-available 1N829A). The Zener itself is good to about 5ppm/C.
Solartron then calibrated each one to find the inflection point on its
voltage/temp curve. The meter has a DAC that sets the current to be at
that point. Then, they add a non-linear temperature-dependent
compensation voltage to eliminate even more tempco. If you look at the
Chinese meter review referred to by other posters, you can see how it
compares to the HP. Pretty impressive.

One more thing, read your cal constants! The critical one is the Zener
bias setting. If you lose the EEPROM, that one would be very difficult
to determine again. The rest can be recovered by a recal.

On 08/04/2011 01:13 PM, volt-nuts-request@febo.com wrote:

Solartron 7081 Test Leads...

--
Bill Ezell

They said 'Windows or better'
so I used Linux.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Bill, Thanks for all the info. I ordered the connectors from Kensington Electronics using the part numbers posed on an earlier email. Very friendly folks but they want a 'company' name. I used my practice. http://www.keiconn.com/ They have a $50 minimum order. One connector and cable clamp meets that requirement. I would appreciate the part number of the switch you ordered. Also, if you could send me a picture of the connector end and the 'probe' end of the cables you found, I would appreciate it. What is the EEPROM part number? I have a couple of programmers that likely will take care of that but I just need to know the EEPROM part number. Also, are there other EPROM's or EEPROM's that need to be archived? Also, are there facilities that still 'calibrate' the 7081? I have been wondering about whether to 'twist' the wires from the connector to the alligator clips, and, if so, which wires to 'twist' and which wires to 'shield'. It might make a difference with 'noise' at least as suggested by one of the service notes I read on the 3458A. Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Bill Ezell Sent: Thursday, August 04, 2011 5:09 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Solartron 7081 Test Leads Joe, If you don't have a cal key, the switch form-factor is standard. I just replaced the switch on my 7081. If I remember correctly, the part number isn't the same, but the mfgr is. I ordered from Digikey. Who did you order the Fischer connectors from? I tracked down part numbers a few years ago, but couldn't find anyone to actually sell them to me. I found one at a flea market, and I did snag two original cables. Regarding the cables/probes, there were several models. The most common have two banana plugs. The connection thru the cable is true 4-wire, terminating at the plug. There was also a version with 5 plugs (4-wire plus shield), and a version with a pair of Kelvin clips (the real ones, one of each pair to each side of the clip). You don't need a Kelvin connection for anything other than resistance measurement. The meter does quite well with the 2-plug version for that anyway. I used 5-wire + shield PTFE / pure copper (picked up on EBay, lots of mil-spec cable shows up there, mostly from aerospace leftovers) terminated with Pomona low-emf plugs. BTW, these are really great meters (except for the read time at 8.5 digits!). Mine keeps cal within typically 1 or 2 ppm for over a year at a time. Yes, seriously. I compare it periodically to my (Fluke-calibrated) pair of Datron 4910 references, and almost never have to recal. (Another wonderful piece of equipment, the 4910). Nice thing about all these old pieces of equipment, they're usually well down the aging curve. :) One thing, they really take 24 hours to fully stabilize, so let it run a few days if you're going to calibrate it. Have you checked out the voltage ref it uses? Amazing.It's just a non-ovenized Zener, albeit one that was designed to be a ref (the still-available 1N829A). The Zener itself is good to about 5ppm/C. Solartron then calibrated each one to find the inflection point on its voltage/temp curve. The meter has a DAC that sets the current to be at that point. Then, they add a non-linear temperature-dependent compensation voltage to eliminate even more tempco. If you look at the Chinese meter review referred to by other posters, you can see how it compares to the HP. Pretty impressive. One more thing, read your cal constants! The critical one is the Zener bias setting. If you lose the EEPROM, that one would be very difficult to determine again. The rest can be recovered by a recal. On 08/04/2011 01:13 PM, volt-nuts-request@febo.com wrote: > Solartron 7081 Test Leads... -- Bill Ezell ---------- They said 'Windows or better' so I used Linux. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
F
Fred
Thu, Aug 4, 2011 11:11 PM

I have the 7061. You let yours on for twenty four hours. I do that with
my calibrators ect but I did it once with my solartron and I allmost had
a fire. It was over 30 degrees Celcius and the meter got so hot it
started smell burned. At 23 degrees, after about an hour my meter gets
towards 50 degrees insite. The transformer around 55 degrees.

Is this normal, is it made to be used in 20 degrees environments or is
there something wronng ? I have no service manual so I can not check
anything. I replaced the nicad, tested the powersupply caps but they
were OK. Could niot find a ripple on the psu that was abnormal (as far
as it is a guess without the manual)

It's to bad because it is a great working meter but as long as I do not
know if this is normal (not allowed to use above certain temps) I do not
use it much.

The cal key locjk is plain standard but I tryed some old keys and after
some modifications it worked :-)

Fred

Bill Ezell schreef op do 04-08-2011 om 18:08 [-0400]:

Joe, If you don't have a cal key, the switch form-factor is standard. I
just replaced the switch on my 7081. If I remember correctly, the part
number isn't the same, but the mfgr is. I ordered from Digikey.

Who did you order the Fischer connectors from? I tracked down part
numbers a few years ago, but couldn't find anyone to actually sell them
to me. I found one at a flea market, and I did snag two original cables.

Regarding the cables/probes, there were several models. The most common
have two banana plugs. The connection thru the cable is true 4-wire,
terminating at the plug. There was also a version with 5 plugs (4-wire
plus shield), and a version with a pair of Kelvin clips (the real ones,
one of each pair to each side of the clip).

You don't need a Kelvin connection for anything other than resistance
measurement. The meter does quite well with the 2-plug version for that
anyway.

I used 5-wire + shield PTFE / pure copper (picked up on EBay, lots of
mil-spec cable shows up there, mostly from aerospace leftovers)
terminated with Pomona low-emf plugs.

BTW, these are really great meters (except for the read time at 8.5
digits!). Mine keeps cal within typically 1 or 2 ppm for over a year at
a time. Yes, seriously. I compare it periodically to my
(Fluke-calibrated) pair of Datron 4910 references, and almost never have
to recal. (Another wonderful piece of equipment, the 4910).
Nice thing about all these old pieces of equipment, they're usually well
down the aging curve. :)

One thing, they really take 24 hours to fully stabilize, so let it run a
few days if you're going to calibrate it.
Have you checked out the voltage ref it uses? Amazing.It's just a
non-ovenized Zener, albeit one that was designed to be a ref (the
still-available 1N829A). The Zener itself is good to about 5ppm/C.
Solartron then calibrated each one to find the inflection point on its
voltage/temp curve. The meter has a DAC that sets the current to be at
that point. Then, they add a non-linear temperature-dependent
compensation voltage to eliminate even more tempco. If you look at the
Chinese meter review referred to by other posters, you can see how it
compares to the HP. Pretty impressive.

One more thing, read your cal constants! The critical one is the Zener
bias setting. If you lose the EEPROM, that one would be very difficult
to determine again. The rest can be recovered by a recal.

On 08/04/2011 01:13 PM, volt-nuts-request@febo.com wrote:

Solartron 7081 Test Leads...

I have the 7061. You let yours on for twenty four hours. I do that with my calibrators ect but I did it once with my solartron and I allmost had a fire. It was over 30 degrees Celcius and the meter got so hot it started smell burned. At 23 degrees, after about an hour my meter gets towards 50 degrees insite. The transformer around 55 degrees. Is this normal, is it made to be used in 20 degrees environments or is there something wronng ? I have no service manual so I can not check anything. I replaced the nicad, tested the powersupply caps but they were OK. Could niot find a ripple on the psu that was abnormal (as far as it is a guess without the manual) It's to bad because it is a great working meter but as long as I do not know if this is normal (not allowed to use above certain temps) I do not use it much. The cal key locjk is plain standard but I tryed some old keys and after some modifications it worked :-) Fred Bill Ezell schreef op do 04-08-2011 om 18:08 [-0400]: > Joe, If you don't have a cal key, the switch form-factor is standard. I > just replaced the switch on my 7081. If I remember correctly, the part > number isn't the same, but the mfgr is. I ordered from Digikey. > > Who did you order the Fischer connectors from? I tracked down part > numbers a few years ago, but couldn't find anyone to actually sell them > to me. I found one at a flea market, and I did snag two original cables. > > Regarding the cables/probes, there were several models. The most common > have two banana plugs. The connection thru the cable is true 4-wire, > terminating at the plug. There was also a version with 5 plugs (4-wire > plus shield), and a version with a pair of Kelvin clips (the real ones, > one of each pair to each side of the clip). > > You don't need a Kelvin connection for anything other than resistance > measurement. The meter does quite well with the 2-plug version for that > anyway. > > I used 5-wire + shield PTFE / pure copper (picked up on EBay, lots of > mil-spec cable shows up there, mostly from aerospace leftovers) > terminated with Pomona low-emf plugs. > > BTW, these are really great meters (except for the read time at 8.5 > digits!). Mine keeps cal within typically 1 or 2 ppm for over a year at > a time. Yes, seriously. I compare it periodically to my > (Fluke-calibrated) pair of Datron 4910 references, and almost never have > to recal. (Another wonderful piece of equipment, the 4910). > Nice thing about all these old pieces of equipment, they're usually well > down the aging curve. :) > > One thing, they really take 24 hours to fully stabilize, so let it run a > few days if you're going to calibrate it. > Have you checked out the voltage ref it uses? Amazing.It's just a > non-ovenized Zener, albeit one that was designed to be a ref (the > still-available 1N829A). The Zener itself is good to about 5ppm/C. > Solartron then calibrated each one to find the inflection point on its > voltage/temp curve. The meter has a DAC that sets the current to be at > that point. Then, they add a non-linear temperature-dependent > compensation voltage to eliminate even more tempco. If you look at the > Chinese meter review referred to by other posters, you can see how it > compares to the HP. Pretty impressive. > > One more thing, read your cal constants! The critical one is the Zener > bias setting. If you lose the EEPROM, that one would be very difficult > to determine again. The rest can be recovered by a recal. > > On 08/04/2011 01:13 PM, volt-nuts-request@febo.com wrote: > > Solartron 7081 Test Leads... >
DC
David C. Partridge
Fri, Aug 5, 2011 2:36 PM

terminated with Pomona low emf plugs

Where did you get those from Bill?  I spoke to Pomona in UK and they won't sell the plugs separately, only as part of their low-EMF lead set.

Regards,
David Partridge

>terminated with Pomona low emf plugs Where did you get those from Bill? I spoke to Pomona in UK and they won't sell the plugs separately, only as part of their low-EMF lead set. Regards, David Partridge