Discussion of precise voltage measurement

View all threads

Voltage standard spotted for sale, 4410

B
ben
Mon, Aug 11, 2014 10:31 AM

Hello all,  spotted for sale on eBay is a Guildline 4410 zener voltage
standard, asking US$995. Outputs are 4 of 10V, and 4 of 1.018V. Comparable
to the Datron 4911. Typical drift I noticed from using another unit 8 years
ago was 1ppm per year. I have a similar unit here manufactured by
Statronics (the OEM for Guildline, I think), however I do not have a
manual sorry.  If anyone does have a manual for this unit I love like a
copy - thanks. Of interest to me; the ebay description describes the 1.018V
C output is non working - which is also the same problem with my unit. I
measured my 1k dividing resistor on my 1.018V C output is open circuit.
Cheerio, ben.

Hello all, spotted for sale on eBay is a Guildline 4410 zener voltage standard, asking US$995. Outputs are 4 of 10V, and 4 of 1.018V. Comparable to the Datron 4911. Typical drift I noticed from using another unit 8 years ago was 1ppm per year. I have a similar unit here manufactured by Statronics (the OEM for Guildline, I think), however I do not have a manual sorry. If anyone does have a manual for this unit I love like a copy - thanks. Of interest to me; the ebay description describes the 1.018V C output is non working - which is also the same problem with my unit. I measured my 1k dividing resistor on my 1.018V C output is open circuit. Cheerio, ben.
W
Will
Mon, Aug 11, 2014 7:24 PM

Comparable to the Datron 4911.

The 4410 is based on ovenised LM329 reference = LM399 without a
heater. You can expect drift by factor of 10 worse than Datron
4910/4911 or a good Fluke 732A.

> Comparable to the Datron 4911. The 4410 is based on ovenised LM329 reference = LM399 without a heater. You can expect drift by factor of 10 worse than Datron 4910/4911 or a good Fluke 732A.
TK
Tom Knox
Mon, Aug 11, 2014 10:25 PM

I am not sure the use of LM329's was a cost saving measure or compromise, actually the 4410 uses a different approch, it's basicly is a big oven. Guildline has been a Metrology leader for decades and  has been a leader including  temp measurements and oil baths so I for one would be interested in how well their approach to a voltage standard performs.
Cheers;
Thomas Knox

Date: Mon, 11 Aug 2014 22:24:57 +0300
From: willvolts@gmail.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Voltage standard spotted for sale, 4410

Comparable to the Datron 4911.

The 4410 is based on ovenised LM329 reference = LM399 without a
heater. You can expect drift by factor of 10 worse than Datron
4910/4911 or a good Fluke 732A.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I am not sure the use of LM329's was a cost saving measure or compromise, actually the 4410 uses a different approch, it's basicly is a big oven. Guildline has been a Metrology leader for decades and has been a leader including temp measurements and oil baths so I for one would be interested in how well their approach to a voltage standard performs. Cheers; Thomas Knox > Date: Mon, 11 Aug 2014 22:24:57 +0300 > From: willvolts@gmail.com > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] Voltage standard spotted for sale, 4410 > > > Comparable to the Datron 4911. > > The 4410 is based on ovenised LM329 reference = LM399 without a > heater. You can expect drift by factor of 10 worse than Datron > 4910/4911 or a good Fluke 732A. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
W
Will
Tue, Aug 12, 2014 8:09 AM

2014-08-12 1:25 UTC+03.00, Tom Knox actast@hotmail.com:

I am not sure the use of LM329's was a cost saving measure or compromise,
actually the 4410 uses a different approch, it's basicly is a big oven.
Guildline has been a Metrology leader for decades and  has been a leader
including  temp measurements and oil baths so I for one would be interested
in how well their approach to a voltage standard performs.

The 4410 was not designed or manufactured by Guildline. Just like many
of the products sold under their brand name.

At the time of design the LTZ1000 was not yet available and the Fluke
ref amp was obviously property of Fluke only. The only same level
technology available to all manufactueres was the 1N829 zener which if
carefully selected and matched can be as good as LTZ1000. Datron
calibrators with multiple selected, aged and matched 1N829 zeners are
a good example of that.

The original 100 page (Statronics) documentation seems to be a masters
thesis rather than a product manual and has a detailed description of
the design process. The text mentions that the 1N829 temperature
compensated zeners were discarded because of the poor availability in
Australia where the unit was developed. The LM399/LM329 series was
selected because it was the "best available". Band-gap type references
were also evaluated but found unsatisfactory at the early stage.

2014-08-12 1:25 UTC+03.00, Tom Knox <actast@hotmail.com>: > I am not sure the use of LM329's was a cost saving measure or compromise, > actually the 4410 uses a different approch, it's basicly is a big oven. > Guildline has been a Metrology leader for decades and has been a leader > including temp measurements and oil baths so I for one would be interested > in how well their approach to a voltage standard performs. The 4410 was not designed or manufactured by Guildline. Just like many of the products sold under their brand name. At the time of design the LTZ1000 was not yet available and the Fluke ref amp was obviously property of Fluke only. The only same level technology available to all manufactueres was the 1N829 zener which if carefully selected and matched can be as good as LTZ1000. Datron calibrators with multiple selected, aged and matched 1N829 zeners are a good example of that. The original 100 page (Statronics) documentation seems to be a masters thesis rather than a product manual and has a detailed description of the design process. The text mentions that the 1N829 temperature compensated zeners were discarded because of the poor availability in Australia where the unit was developed. The LM399/LM329 series was selected because it was the "best available". Band-gap type references were also evaluated but found unsatisfactory at the early stage.
RE
roman's ebay
Wed, Aug 13, 2014 7:26 AM

Hi All
Newcomer to the list. Been watching it since I came across it about a month
ago. Didn't realise that I wasn't the ONLY person in the world obsessed with
having an accurate volt in my back yard for no reason other than that I
can.....:-)

This discussion about the Statronics VS4 particularly got my interest. I
have one courtesy of fleabay about 4 years ago, for the princely sum of $20
plus delivery (mine was the only bid, which (together with the quad Standard
Cell box for about the same) is why I thought I was alone...).  The VS4 (and
indeed the Standard Cell Box) appears to work very well and all outputs have
maintained their relativity over that time within the limits of my
measurement patience, even after about a year powered down after I moved
house. Plan to start doing some serious tracking real soon now.

Each of the 4 independent cells in the VS4 consist of 8 selected, matched
and aged LM329s with matched resistors (all off the same reel of Evanohm,
selected mica cards, etc), low drift amps etc all contained in a controlled
oven. The Statronics stability spec is -2 +-2ppm per year (after year one).
The unit was conceived and designed by the CSIRO who run (or at least ran)
the Australian National Measurement Laboratory, so I would guess that the
spec is conservative. If Will is correct about  the Gould 4410/Fluke 732A,
these must be pretty good devices. I suspect I won't get one of them for
$20!

I saw Ben's question about the manual, and Tom's response about the 100 page
masters thesis. That would be interesting to read!  I have a copy of the
Statronics VS4 Instruction Manual Issue 2 which is only about 70 pages, half
of which is a treatise on statistical intercomparisons and a HP-85 Basic
program. But the first 38 pages contain what I would expect - user
instructions, design descriptions, schematics, etc. Unfortunately, I don't
have an electronic copy, nor ready access to a scanner.

Ben, I note you are in Australia. Whereabouts? I am in Adelaide. Maybe I can
get the hard copy to you some way.

Roman

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Will
Sent: Tuesday, 12 August 2014 5:40 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Voltage standard spotted for sale, 4410

2014-08-12 1:25 UTC+03.00, Tom Knox actast@hotmail.com:

I am not sure the use of LM329's was a cost saving measure or
compromise, actually the 4410 uses a different approch, it's basicly is a

big oven.

Guildline has been a Metrology leader for decades and  has been a
leader including  temp measurements and oil baths so I for one would
be interested in how well their approach to a voltage standard performs.

The 4410 was not designed or manufactured by Guildline. Just like many of
the products sold under their brand name.

At the time of design the LTZ1000 was not yet available and the Fluke ref
amp was obviously property of Fluke only. The only same level technology
available to all manufactueres was the 1N829 zener which if carefully
selected and matched can be as good as LTZ1000. Datron calibrators with
multiple selected, aged and matched 1N829 zeners are a good example of that.

The original 100 page (Statronics) documentation seems to be a masters
thesis rather than a product manual and has a detailed description of the
design process. The text mentions that the 1N829 temperature compensated
zeners were discarded because of the poor availability in Australia where
the unit was developed. The LM399/LM329 series was selected because it was
the "best available". Band-gap type references were also evaluated but found
unsatisfactory at the early stage.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi All Newcomer to the list. Been watching it since I came across it about a month ago. Didn't realise that I wasn't the ONLY person in the world obsessed with having an accurate volt in my back yard for no reason other than that I can.....:-) This discussion about the Statronics VS4 particularly got my interest. I have one courtesy of fleabay about 4 years ago, for the princely sum of $20 plus delivery (mine was the only bid, which (together with the quad Standard Cell box for about the same) is why I thought I was alone...). The VS4 (and indeed the Standard Cell Box) appears to work very well and all outputs have maintained their relativity over that time within the limits of my measurement patience, even after about a year powered down after I moved house. Plan to start doing some serious tracking real soon now. Each of the 4 independent cells in the VS4 consist of 8 selected, matched and aged LM329s with matched resistors (all off the same reel of Evanohm, selected mica cards, etc), low drift amps etc all contained in a controlled oven. The Statronics stability spec is -2 +-2ppm per year (after year one). The unit was conceived and designed by the CSIRO who run (or at least ran) the Australian National Measurement Laboratory, so I would guess that the spec is conservative. If Will is correct about the Gould 4410/Fluke 732A, these must be pretty good devices. I suspect I won't get one of them for $20! I saw Ben's question about the manual, and Tom's response about the 100 page masters thesis. That would be interesting to read! I have a copy of the Statronics VS4 Instruction Manual Issue 2 which is only about 70 pages, half of which is a treatise on statistical intercomparisons and a HP-85 Basic program. But the first 38 pages contain what I would expect - user instructions, design descriptions, schematics, etc. Unfortunately, I don't have an electronic copy, nor ready access to a scanner. Ben, I note you are in Australia. Whereabouts? I am in Adelaide. Maybe I can get the hard copy to you some way. Roman -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Will Sent: Tuesday, 12 August 2014 5:40 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Voltage standard spotted for sale, 4410 2014-08-12 1:25 UTC+03.00, Tom Knox <actast@hotmail.com>: > I am not sure the use of LM329's was a cost saving measure or > compromise, actually the 4410 uses a different approch, it's basicly is a big oven. > Guildline has been a Metrology leader for decades and has been a > leader including temp measurements and oil baths so I for one would > be interested in how well their approach to a voltage standard performs. The 4410 was not designed or manufactured by Guildline. Just like many of the products sold under their brand name. At the time of design the LTZ1000 was not yet available and the Fluke ref amp was obviously property of Fluke only. The only same level technology available to all manufactueres was the 1N829 zener which if carefully selected and matched can be as good as LTZ1000. Datron calibrators with multiple selected, aged and matched 1N829 zeners are a good example of that. The original 100 page (Statronics) documentation seems to be a masters thesis rather than a product manual and has a detailed description of the design process. The text mentions that the 1N829 temperature compensated zeners were discarded because of the poor availability in Australia where the unit was developed. The LM399/LM329 series was selected because it was the "best available". Band-gap type references were also evaluated but found unsatisfactory at the early stage. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
JW
Jason Watson
Wed, Aug 13, 2014 5:14 PM

The Guildline 4410 technical manual can now be found on Didier's website (
http://www.ko4bb.com/manuals/).  Look in the Recent Uploads folder until
Didier gets it moved around.

On Wed, Aug 13, 2014 at 3:26 AM, roman's ebay romeo987@westnet.com.au
wrote:

Hi All
Newcomer to the list. Been watching it since I came across it about a month
ago. Didn't realise that I wasn't the ONLY person in the world obsessed
with
having an accurate volt in my back yard for no reason other than that I
can.....:-)

This discussion about the Statronics VS4 particularly got my interest. I
have one courtesy of fleabay about 4 years ago, for the princely sum of $20
plus delivery (mine was the only bid, which (together with the quad
Standard
Cell box for about the same) is why I thought I was alone...).  The VS4
(and
indeed the Standard Cell Box) appears to work very well and all outputs
have
maintained their relativity over that time within the limits of my
measurement patience, even after about a year powered down after I moved
house. Plan to start doing some serious tracking real soon now.

Each of the 4 independent cells in the VS4 consist of 8 selected, matched
and aged LM329s with matched resistors (all off the same reel of Evanohm,
selected mica cards, etc), low drift amps etc all contained in a controlled
oven. The Statronics stability spec is -2 +-2ppm per year (after year one).
The unit was conceived and designed by the CSIRO who run (or at least ran)
the Australian National Measurement Laboratory, so I would guess that the
spec is conservative. If Will is correct about  the Gould 4410/Fluke 732A,
these must be pretty good devices. I suspect I won't get one of them for
$20!

I saw Ben's question about the manual, and Tom's response about the 100
page
masters thesis. That would be interesting to read!  I have a copy of the
Statronics VS4 Instruction Manual Issue 2 which is only about 70 pages,
half
of which is a treatise on statistical intercomparisons and a HP-85 Basic
program. But the first 38 pages contain what I would expect - user
instructions, design descriptions, schematics, etc. Unfortunately, I don't
have an electronic copy, nor ready access to a scanner.

Ben, I note you are in Australia. Whereabouts? I am in Adelaide. Maybe I
can
get the hard copy to you some way.

Roman

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Will
Sent: Tuesday, 12 August 2014 5:40 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Voltage standard spotted for sale, 4410

2014-08-12 1:25 UTC+03.00, Tom Knox actast@hotmail.com:

I am not sure the use of LM329's was a cost saving measure or
compromise, actually the 4410 uses a different approch, it's basicly is a

big oven.

Guildline has been a Metrology leader for decades and  has been a
leader including  temp measurements and oil baths so I for one would
be interested in how well their approach to a voltage standard performs.

The 4410 was not designed or manufactured by Guildline. Just like many of
the products sold under their brand name.

At the time of design the LTZ1000 was not yet available and the Fluke ref
amp was obviously property of Fluke only. The only same level technology
available to all manufactueres was the 1N829 zener which if carefully
selected and matched can be as good as LTZ1000. Datron calibrators with
multiple selected, aged and matched 1N829 zeners are a good example of
that.

The original 100 page (Statronics) documentation seems to be a masters
thesis rather than a product manual and has a detailed description of the
design process. The text mentions that the 1N829 temperature compensated
zeners were discarded because of the poor availability in Australia where
the unit was developed. The LM399/LM329 series was selected because it was
the "best available". Band-gap type references were also evaluated but
found
unsatisfactory at the early stage.


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

The Guildline 4410 technical manual can now be found on Didier's website ( http://www.ko4bb.com/manuals/). Look in the Recent Uploads folder until Didier gets it moved around. On Wed, Aug 13, 2014 at 3:26 AM, roman's ebay <romeo987@westnet.com.au> wrote: > Hi All > Newcomer to the list. Been watching it since I came across it about a month > ago. Didn't realise that I wasn't the ONLY person in the world obsessed > with > having an accurate volt in my back yard for no reason other than that I > can.....:-) > > This discussion about the Statronics VS4 particularly got my interest. I > have one courtesy of fleabay about 4 years ago, for the princely sum of $20 > plus delivery (mine was the only bid, which (together with the quad > Standard > Cell box for about the same) is why I thought I was alone...). The VS4 > (and > indeed the Standard Cell Box) appears to work very well and all outputs > have > maintained their relativity over that time within the limits of my > measurement patience, even after about a year powered down after I moved > house. Plan to start doing some serious tracking real soon now. > > Each of the 4 independent cells in the VS4 consist of 8 selected, matched > and aged LM329s with matched resistors (all off the same reel of Evanohm, > selected mica cards, etc), low drift amps etc all contained in a controlled > oven. The Statronics stability spec is -2 +-2ppm per year (after year one). > The unit was conceived and designed by the CSIRO who run (or at least ran) > the Australian National Measurement Laboratory, so I would guess that the > spec is conservative. If Will is correct about the Gould 4410/Fluke 732A, > these must be pretty good devices. I suspect I won't get one of them for > $20! > > I saw Ben's question about the manual, and Tom's response about the 100 > page > masters thesis. That would be interesting to read! I have a copy of the > Statronics VS4 Instruction Manual Issue 2 which is only about 70 pages, > half > of which is a treatise on statistical intercomparisons and a HP-85 Basic > program. But the first 38 pages contain what I would expect - user > instructions, design descriptions, schematics, etc. Unfortunately, I don't > have an electronic copy, nor ready access to a scanner. > > Ben, I note you are in Australia. Whereabouts? I am in Adelaide. Maybe I > can > get the hard copy to you some way. > > Roman > > > > -----Original Message----- > From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On > Behalf Of Will > Sent: Tuesday, 12 August 2014 5:40 PM > To: Discussion of precise voltage measurement > Subject: Re: [volt-nuts] Voltage standard spotted for sale, 4410 > > 2014-08-12 1:25 UTC+03.00, Tom Knox <actast@hotmail.com>: > > I am not sure the use of LM329's was a cost saving measure or > > compromise, actually the 4410 uses a different approch, it's basicly is a > big oven. > > Guildline has been a Metrology leader for decades and has been a > > leader including temp measurements and oil baths so I for one would > > be interested in how well their approach to a voltage standard performs. > > The 4410 was not designed or manufactured by Guildline. Just like many of > the products sold under their brand name. > > At the time of design the LTZ1000 was not yet available and the Fluke ref > amp was obviously property of Fluke only. The only same level technology > available to all manufactueres was the 1N829 zener which if carefully > selected and matched can be as good as LTZ1000. Datron calibrators with > multiple selected, aged and matched 1N829 zeners are a good example of > that. > > The original 100 page (Statronics) documentation seems to be a masters > thesis rather than a product manual and has a detailed description of the > design process. The text mentions that the 1N829 temperature compensated > zeners were discarded because of the poor availability in Australia where > the unit was developed. The LM399/LM329 series was selected because it was > the "best available". Band-gap type references were also evaluated but > found > unsatisfactory at the early stage. > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >