W
WarrenS
Sat, Sep 3, 2011 10:34 PM
Want a stable 10V ref?
Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
BUT
For the Extreme (or cheap) Volt Nut, There is another choice.
Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
It has ONLY ONE critical part that limits its accuracy and stability.
Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
The main advantage of all this extra trouble is low cost accuracy and stability.
There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
Nothing very special is needed to build this beside care and time.
To get the best performance, It does take a lot of Nut-time to match and select things.
Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
Any good 6V type reference device could be used.
I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
ws
Want a stable 10V ref?
Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
BUT
For the Extreme (or cheap) Volt Nut, There is another choice.
Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
It has ONLY ONE critical part that limits its accuracy and stability.
Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
The main advantage of all this extra trouble is low cost accuracy and stability.
There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
Nothing very special is needed to build this beside care and time.
To get the best performance, It does take a lot of Nut-time to match and select things.
Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
Any good 6V type reference device could be used.
I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
ws
W
WB6BNQ
Sat, Sep 3, 2011 11:59 PM
Warren,
You need to do your schematics a little bit better. Besides, I think you attached the wrong one. There are a number of errors on the attached schematic.
Bill....WB6BNQ
WarrenS wrote:
Want a stable 10V ref?
Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
BUT
For the Extreme (or cheap) Volt Nut, There is another choice.
Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
It has ONLY ONE critical part that limits its accuracy and stability.
Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
The main advantage of all this extra trouble is low cost accuracy and stability.
There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
Nothing very special is needed to build this beside care and time.
To get the best performance, It does take a lot of Nut-time to match and select things.
Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
Any good 6V type reference device could be used.
I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
ws
Name: ws10Vref-9-3LR.gif
ws10Vref-9-3LR.gif Type: GIF Image (image/gif)
Encoding: base64
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Warren,
You need to do your schematics a little bit better. Besides, I think you attached the wrong one. There are a number of errors on the attached schematic.
Bill....WB6BNQ
WarrenS wrote:
> Want a stable 10V ref?
> Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
> BUT
> For the Extreme (or cheap) Volt Nut, There is another choice.
> Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
>
> It has ONLY ONE critical part that limits its accuracy and stability.
> Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
>
> Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
> To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
> This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
> Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
> After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
>
> The main advantage of all this extra trouble is low cost accuracy and stability.
> There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
>
> Nothing very special is needed to build this beside care and time.
> To get the best performance, It does take a lot of Nut-time to match and select things.
> Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
>
> Any good 6V type reference device could be used.
> I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
> I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
> and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
>
> ws
>
> ------------------------------------------------------------------------
> Name: ws10Vref-9-3LR.gif
> ws10Vref-9-3LR.gif Type: GIF Image (image/gif)
> Encoding: base64
>
> ------------------------------------------------------------------------
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
JD
John Devereux
Sun, Sep 4, 2011 2:56 PM
Hi Bill, Warren,
Whats the matter with it? The schematic is clear enough to me,
interesting and seems quite clever to this wannabe nut.
If I understand correctly, the opamp adjusts the divided-down output via
the regulator so that it matches the reference. Everything is powered
from the 10V output, eliminating problems with variable input voltages
and opamp PSRR.
I would be interested to know:
-
Any suggestions for the wirewound resistors? Vishay bulk metal foil
resistors are not cheap but say 5ppm wirewounds are not pennies
either, I don't know how many you would need before a sufficient
number could be matched.
-
I am unclear as to the precise function of the 160k and 40M
resistors. I suppose one of them shifts the output, and the other
compensates the zener so it still runs at the same minimum-tempco
current?
The zeners do look interesting as references, I admit I have been
fixated on LTZ1000s.
Farnell/Newark don't stock the 1N825 but they do stock the
1N825A.
http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825
According to its "datasheet"
http://www.farnell.com/datasheets/332215.pdf there are also some
similar looking parts that may have have improved tempco, for example
1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
are even some 2ppm/K parts but these seem unobtainium.
Are there any better datasheets? For example how does the tempco itself
vary with temperature? The company does not seem to have a web presence.
There
Regards,
John
WB6BNQ wb6bnq@cox.net writes:
Warren,
You need to do your schematics a little bit better. Besides, I think you attached the wrong one. There are a number of errors on the attached schematic.
Bill....WB6BNQ
WarrenS wrote:
Want a stable 10V ref?
Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
BUT
For the Extreme (or cheap) Volt Nut, There is another choice.
Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
It has ONLY ONE critical part that limits its accuracy and stability.
Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
The main advantage of all this extra trouble is low cost accuracy and stability.
There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
Nothing very special is needed to build this beside care and time.
To get the best performance, It does take a lot of Nut-time to match and select things.
Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
Any good 6V type reference device could be used.
I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
Hi Bill, Warren,
Whats the matter with it? The schematic is clear enough to me,
interesting and seems quite clever to this wannabe nut.
If I understand correctly, the opamp adjusts the divided-down output via
the regulator so that it matches the reference. Everything is powered
from the 10V output, eliminating problems with variable input voltages
and opamp PSRR.
I would be interested to know:
- Any suggestions for the wirewound resistors? Vishay bulk metal foil
resistors are not cheap but say 5ppm wirewounds are not pennies
either, I don't know how many you would need before a sufficient
number could be matched.
- I am unclear as to the precise function of the 160k and 40M
resistors. I suppose one of them shifts the output, and the other
compensates the zener so it still runs at the same minimum-tempco
current?
The zeners do look interesting as references, I admit I have been
fixated on LTZ1000s.
Farnell/Newark don't stock the 1N825 but they do stock the
1N825A.
<http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825>
According to its "datasheet"
<http://www.farnell.com/datasheets/332215.pdf> there are also some
similar looking parts that may have have improved tempco, for example
1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
are even some 2ppm/K parts but these seem unobtainium.
Are there any better datasheets? For example how does the tempco itself
vary with temperature? The company does not seem to have a web presence.
There
Regards,
John
WB6BNQ <wb6bnq@cox.net> writes:
> Warren,
>
> You need to do your schematics a little bit better. Besides, I think you attached the wrong one. There are a number of errors on the attached schematic.
>
> Bill....WB6BNQ
>
> WarrenS wrote:
>
>> Want a stable 10V ref?
>> Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
>> BUT
>> For the Extreme (or cheap) Volt Nut, There is another choice.
>> Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
>>
>> It has ONLY ONE critical part that limits its accuracy and stability.
>> Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
>>
>> Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
>> To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
>> This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
>> Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
>> After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
>>
>> The main advantage of all this extra trouble is low cost accuracy and stability.
>> There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
>>
>> Nothing very special is needed to build this beside care and time.
>> To get the best performance, It does take a lot of Nut-time to match and select things.
>> Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
>>
>> Any good 6V type reference device could be used.
>> I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
>> I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
>> and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
--
John Devereux
MK
m k
Sun, Sep 4, 2011 4:28 PM
Hi,
The schematic has the opamp+ connected to the wrong point and there are some other details wrong. In another job there was an instrument that used an 1n829a run at about 7.5mA for a 10 V reference that used to hold to better than 1uV over lab temperatures. It was done using the Op amp to generate the 10V that directly drove the 7.5mA into the zener, select on test resistors. but for 0.1ppm you need to address opamp heating and unintended thermocouples everywhere.
From: john@devereux.me.uk
To: volt-nuts@febo.com
Date: Sun, 4 Sep 2011 15:56:00 +0100
Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
Hi Bill, Warren,
Whats the matter with it? The schematic is clear enough to me,
interesting and seems quite clever to this wannabe nut.
If I understand correctly, the opamp adjusts the divided-down output via
the regulator so that it matches the reference. Everything is powered
from the 10V output, eliminating problems with variable input voltages
and opamp PSRR.
I would be interested to know:
-
Any suggestions for the wirewound resistors? Vishay bulk metal foil
resistors are not cheap but say 5ppm wirewounds are not pennies
either, I don't know how many you would need before a sufficient
number could be matched.
-
I am unclear as to the precise function of the 160k and 40M
resistors. I suppose one of them shifts the output, and the other
compensates the zener so it still runs at the same minimum-tempco
current?
The zeners do look interesting as references, I admit I have been
fixated on LTZ1000s.
Farnell/Newark don't stock the 1N825 but they do stock the
1N825A.
http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825
According to its "datasheet"
http://www.farnell.com/datasheets/332215.pdf there are also some
similar looking parts that may have have improved tempco, for example
1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
are even some 2ppm/K parts but these seem unobtainium.
Are there any better datasheets? For example how does the tempco itself
vary with temperature? The company does not seem to have a web presence.
There
Regards,
John
WB6BNQ wb6bnq@cox.net writes:
Warren,
You need to do your schematics a little bit better. Besides, I think you attached the wrong one. There are a number of errors on the attached schematic.
Bill....WB6BNQ
WarrenS wrote:
Want a stable 10V ref?
Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
BUT
For the Extreme (or cheap) Volt Nut, There is another choice.
Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
It has ONLY ONE critical part that limits its accuracy and stability.
Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
The main advantage of all this extra trouble is low cost accuracy and stability.
There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
Nothing very special is needed to build this beside care and time.
To get the best performance, It does take a lot of Nut-time to match and select things.
Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
Any good 6V type reference device could be used.
I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
Hi,
The schematic has the opamp+ connected to the wrong point and there are some other details wrong. In another job there was an instrument that used an 1n829a run at about 7.5mA for a 10 V reference that used to hold to better than 1uV over lab temperatures. It was done using the Op amp to generate the 10V that directly drove the 7.5mA into the zener, select on test resistors. but for 0.1ppm you need to address opamp heating and unintended thermocouples everywhere.
> From: john@devereux.me.uk
> To: volt-nuts@febo.com
> Date: Sun, 4 Sep 2011 15:56:00 +0100
> Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
>
>
> Hi Bill, Warren,
>
> Whats the matter with it? The schematic is clear enough to me,
> interesting and seems quite clever to this wannabe nut.
>
> If I understand correctly, the opamp adjusts the divided-down output via
> the regulator so that it matches the reference. Everything is powered
> from the 10V output, eliminating problems with variable input voltages
> and opamp PSRR.
>
> I would be interested to know:
>
> - Any suggestions for the wirewound resistors? Vishay bulk metal foil
> resistors are not cheap but say 5ppm wirewounds are not pennies
> either, I don't know how many you would need before a sufficient
> number could be matched.
>
> - I am unclear as to the precise function of the 160k and 40M
> resistors. I suppose one of them shifts the output, and the other
> compensates the zener so it still runs at the same minimum-tempco
> current?
>
> The zeners do look interesting as references, I admit I have been
> fixated on LTZ1000s.
>
> Farnell/Newark don't stock the 1N825 but they do stock the
> 1N825A.
>
> <http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825>
>
> According to its "datasheet"
> <http://www.farnell.com/datasheets/332215.pdf> there are also some
> similar looking parts that may have have improved tempco, for example
> 1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
> are even some 2ppm/K parts but these seem unobtainium.
>
> Are there any better datasheets? For example how does the tempco itself
> vary with temperature? The company does not seem to have a web presence.
>
> There
>
> Regards,
>
> John
>
>
> WB6BNQ <wb6bnq@cox.net> writes:
>
> > Warren,
> >
> > You need to do your schematics a little bit better. Besides, I think you attached the wrong one. There are a number of errors on the attached schematic.
> >
> > Bill....WB6BNQ
> >
> > WarrenS wrote:
> >
> >> Want a stable 10V ref?
> >> Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
> >> BUT
> >> For the Extreme (or cheap) Volt Nut, There is another choice.
> >> Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
> >>
> >> It has ONLY ONE critical part that limits its accuracy and stability.
> >> Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
> >>
> >> Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
> >> To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
> >> This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
> >> Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
> >> After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
> >>
> >> The main advantage of all this extra trouble is low cost accuracy and stability.
> >> There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
> >>
> >> Nothing very special is needed to build this beside care and time.
> >> To get the best performance, It does take a lot of Nut-time to match and select things.
> >> Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
> >>
> >> Any good 6V type reference device could be used.
> >> I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
> >> I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
> >> and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
>
> --
>
> John Devereux
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
FS
Fred Schneider
Sun, Sep 4, 2011 4:50 PM
Why is that ? I think it is correct. The opamp wants both its inputs at 6,2 V so he keeps his output around 5 V, so the output of the 7805 will be 10V. If the output drops the inverting input sees a lower voltage at the divider compared to the reference and the output raises until both are equal. The opamp is fed by the 10 V coming out the 7805. So far from its rail. I think it is a nice idea. But maybe I take a TL431 instead of the 7805 ( if i remember well that can deliver 100 mA and a better opamp like the OPA277. )
I think I gonna try one.
If I'm not wrong Fluke does something like this in the 731A but I have the schematic not on hand right now.
Fred PA4TIM
Op 4 sep. 2011 om 18:28 heeft m k m1k3k1@hotmail.com het volgende geschreven:
Hi,
The schematic has the opamp+ connected to the wrong point and there are some other details wrong. In another job there was an instrument that used an 1n829a run at about 7.5mA for a 10 V reference that used to hold to better than 1uV over lab temperatures. It was done using the Op amp to generate the 10V that directly drove the 7.5mA into the zener, select on test resistors. but for 0.1ppm you need to address opamp heating and unintended thermocouples everywhere.
From: john@devereux.me.uk
To: volt-nuts@febo.com
Date: Sun, 4 Sep 2011 15:56:00 +0100
Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
Hi Bill, Warren,
Whats the matter with it? The schematic is clear enough to me,
interesting and seems quite clever to this wannabe nut.
If I understand correctly, the opamp adjusts the divided-down output via
the regulator so that it matches the reference. Everything is powered
from the 10V output, eliminating problems with variable input voltages
and opamp PSRR.
I would be interested to know:
-
Any suggestions for the wirewound resistors? Vishay bulk metal foil
resistors are not cheap but say 5ppm wirewounds are not pennies
either, I don't know how many you would need before a sufficient
number could be matched.
-
I am unclear as to the precise function of the 160k and 40M
resistors. I suppose one of them shifts the output, and the other
compensates the zener so it still runs at the same minimum-tempco
current?
The zeners do look interesting as references, I admit I have been
fixated on LTZ1000s.
Farnell/Newark don't stock the 1N825 but they do stock the
1N825A.
http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825
According to its "datasheet"
http://www.farnell.com/datasheets/332215.pdf there are also some
similar looking parts that may have have improved tempco, for example
1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
are even some 2ppm/K parts but these seem unobtainium.
Are there any better datasheets? For example how does the tempco itself
vary with temperature? The company does not seem to have a web presence.
There
Regards,
John
WB6BNQ wb6bnq@cox.net writes:
Warren,
You need to do your schematics a little bit better. Besides, I think you attached the wrong one. There are a number of errors on the attached schematic.
Bill....WB6BNQ
WarrenS wrote:
Want a stable 10V ref?
Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
BUT
For the Extreme (or cheap) Volt Nut, There is another choice.
Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
It has ONLY ONE critical part that limits its accuracy and stability.
Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
The main advantage of all this extra trouble is low cost accuracy and stability.
There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
Nothing very special is needed to build this beside care and time.
To get the best performance, It does take a lot of Nut-time to match and select things.
Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
Any good 6V type reference device could be used.
I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
Why is that ? I think it is correct. The opamp wants both its inputs at 6,2 V so he keeps his output around 5 V, so the output of the 7805 will be 10V. If the output drops the inverting input sees a lower voltage at the divider compared to the reference and the output raises until both are equal. The opamp is fed by the 10 V coming out the 7805. So far from its rail. I think it is a nice idea. But maybe I take a TL431 instead of the 7805 ( if i remember well that can deliver 100 mA and a better opamp like the OPA277. )
I think I gonna try one.
If I'm not wrong Fluke does something like this in the 731A but I have the schematic not on hand right now.
Fred PA4TIM
Op 4 sep. 2011 om 18:28 heeft m k <m1k3k1@hotmail.com> het volgende geschreven:
>
> Hi,
> The schematic has the opamp+ connected to the wrong point and there are some other details wrong. In another job there was an instrument that used an 1n829a run at about 7.5mA for a 10 V reference that used to hold to better than 1uV over lab temperatures. It was done using the Op amp to generate the 10V that directly drove the 7.5mA into the zener, select on test resistors. but for 0.1ppm you need to address opamp heating and unintended thermocouples everywhere.
>
>> From: john@devereux.me.uk
>> To: volt-nuts@febo.com
>> Date: Sun, 4 Sep 2011 15:56:00 +0100
>> Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
>>
>>
>> Hi Bill, Warren,
>>
>> Whats the matter with it? The schematic is clear enough to me,
>> interesting and seems quite clever to this wannabe nut.
>>
>> If I understand correctly, the opamp adjusts the divided-down output via
>> the regulator so that it matches the reference. Everything is powered
>> from the 10V output, eliminating problems with variable input voltages
>> and opamp PSRR.
>>
>> I would be interested to know:
>>
>> - Any suggestions for the wirewound resistors? Vishay bulk metal foil
>> resistors are not cheap but say 5ppm wirewounds are not pennies
>> either, I don't know how many you would need before a sufficient
>> number could be matched.
>>
>> - I am unclear as to the precise function of the 160k and 40M
>> resistors. I suppose one of them shifts the output, and the other
>> compensates the zener so it still runs at the same minimum-tempco
>> current?
>>
>> The zeners do look interesting as references, I admit I have been
>> fixated on LTZ1000s.
>>
>> Farnell/Newark don't stock the 1N825 but they do stock the
>> 1N825A.
>>
>> <http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825>
>>
>> According to its "datasheet"
>> <http://www.farnell.com/datasheets/332215.pdf> there are also some
>> similar looking parts that may have have improved tempco, for example
>> 1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
>> are even some 2ppm/K parts but these seem unobtainium.
>>
>> Are there any better datasheets? For example how does the tempco itself
>> vary with temperature? The company does not seem to have a web presence.
>>
>> There
>>
>> Regards,
>>
>> John
>>
>>
>> WB6BNQ <wb6bnq@cox.net> writes:
>>
>>> Warren,
>>>
>>> You need to do your schematics a little bit better. Besides, I think you attached the wrong one. There are a number of errors on the attached schematic.
>>>
>>> Bill....WB6BNQ
>>>
>>> WarrenS wrote:
>>>
>>>> Want a stable 10V ref?
>>>> Best thing to do is get a fluke 732A/B and have it calibrated regularly for a few years.
>>>> BUT
>>>> For the Extreme (or cheap) Volt Nut, There is another choice.
>>>> Attached is the circuit of a 0.1 PPM, 10 volt transfer standard.
>>>>
>>>> It has ONLY ONE critical part that limits its accuracy and stability.
>>>> Everything else can be made to have so little effect as to be insignificant or else can be trimmed and calibrated out without using a precision voltage source using a method similar to what the daily auto cal in the HP3458 does. It measures it's own internal 6+V reference.
>>>>
>>>> Every year or more, best to record the value of the nominal 6.2 volts, to measurement the ageing rate, by using a known external 10 volt standard.
>>>> To do that, an accurate external way is needed to find the exact ratio of the 6.2 volts to the external 10 volt standard.
>>>> This can be done by using a high resolution Linear DMM (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null meter (aka Fluke 845).
>>>> Lets assume if one wants to do 0.1 PPM things, then they have the equipment and skill needed to measure a 10v to 6v ratio to 0.1 PPM.
>>>> After that ratio is known and recorded the units internal 10 volts can then be set to that same ratio without further need of the external 10V standard.
>>>>
>>>> The main advantage of all this extra trouble is low cost accuracy and stability.
>>>> There are no critical or expensive resistor needed or anything else critical, because all the TC and long term drift are canceled out whenever a manual trim is done using an external ratio device to check its own 10 V against it's own 6.2 volts.
>>>>
>>>> Nothing very special is needed to build this beside care and time.
>>>> To get the best performance, It does take a lot of Nut-time to match and select things.
>>>> Total parts cost with a well supplied analog junk box (or EBay) can be under $10.
>>>>
>>>> Any good 6V type reference device could be used.
>>>> I'm using a well aged and selected 1n825 because I have a lot of history with them and they work good without an oven.
>>>> I have not done any mailing test on these parts but they can be Zero TCed so that they do not change 0.1PPM over normal room temperature
>>>> and they show no change when powered down for short periods of time and tend to return to their original value when hit with a heat gun and cold stray, so it makes a good part to start experimenting with.
>>
>> --
>>
>> John Devereux
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
PK
Poul-Henning Kamp
Sun, Sep 4, 2011 5:05 PM
But for 0.1ppm you need to address opamp heating and unintended
thermocouples everywhere.
Actually, if you go that far anyway, adding a peltier and ntc to
hold the temperature constant is not a big deal.
Build the PCB (both Vref & Tctl) as surface-mount with no through-hole
and attach it directly to the Peltier with a thermal pad
The Peltier needs very little power if the temperature difference
is on the order of a few degrees and the power dissipation in the
controlled volume is low.
When I ran the MAX6350 experiment, the 12V Peltier was running
at slightly less than 1V most of the time.
I would let the Tctl output an analog control voltage and put the
Peltier power-driver outside the controlled volume.
I would recommend preregulating the input power with a linear
regulator (also outside the controlled volume) so that the power
dissipation in the controlled volume is as constant as possible.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
In message <BLU160-W589E11CA52703B9DEA19E7911A0@phx.gbl>, m k writes:
>But for 0.1ppm you need to address opamp heating and unintended
>thermocouples everywhere.
Actually, if you go that far anyway, adding a peltier and ntc to
hold the temperature constant is not a big deal.
Build the PCB (both Vref & Tctl) as surface-mount with no through-hole
and attach it directly to the Peltier with a thermal pad
The Peltier needs very little power if the temperature difference
is on the order of a few degrees and the power dissipation in the
controlled volume is low.
When I ran the MAX6350 experiment, the 12V Peltier was running
at slightly less than 1V most of the time.
I would let the Tctl output an analog control voltage and put the
Peltier power-driver outside the controlled volume.
I would recommend preregulating the input power with a linear
regulator (also outside the controlled volume) so that the power
dissipation in the controlled volume is as constant as possible.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
JD
John Devereux
Sun, Sep 4, 2011 6:39 PM
Hi, The schematic has the opamp+ connected to the wrong point and
there are some other details wrong.
Do you mean the +input or the +supply? They both look OK to me but noone
is saying what the problem is!
In another job there was an instrument that used an 1n829a run at
about 7.5mA for a 10 V reference that used to hold to better than 1uV
over lab temperatures. It was done using the Op amp to generate the
10V that directly drove the 7.5mA into the zener, select on test
resistors. but for 0.1ppm you need to address opamp heating and
unintended thermocouples everywhere.
This is essentially what he is doing as far as I can see. But the opamp
heating is addressed by using the 78L05 as a "power driver", so the
opamp generates no heat! The heat-generating part is taken off-chip (or
off-board perhaps) so it does not affect the input offset voltage.
I agree 0.1ppm seems optimistic, but then I haven't built it. In my
admittedly limited experience thermocouples are not such a problem if
you keep everything isothermal, i.e. avoid temperature
gradients. Stating the obvious perhaps.
Even if it were 1ppm instead 0.1ppm I think this is still an order of
magnitude better than the homebrew ideas discussed so far (if
achievable).
John
From: john@devereux.me.uk
To: volt-nuts@febo.com
Date: Sun, 4 Sep 2011 15:56:00 +0100
Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
Hi Bill, Warren,
Whats the matter with it? The schematic is clear enough to me,
interesting and seems quite clever to this wannabe nut.
If I understand correctly, the opamp adjusts the divided-down output via
the regulator so that it matches the reference. Everything is powered
from the 10V output, eliminating problems with variable input voltages
and opamp PSRR.
I would be interested to know:
-
Any suggestions for the wirewound resistors? Vishay bulk metal foil
resistors are not cheap but say 5ppm wirewounds are not pennies
either, I don't know how many you would need before a sufficient
number could be matched.
-
I am unclear as to the precise function of the 160k and 40M
resistors. I suppose one of them shifts the output, and the other
compensates the zener so it still runs at the same minimum-tempco
current?
The zeners do look interesting as references, I admit I have been
fixated on LTZ1000s.
Farnell/Newark don't stock the 1N825 but they do stock the
1N825A.
http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825
According to its "datasheet"
http://www.farnell.com/datasheets/332215.pdf there are also some
similar looking parts that may have have improved tempco, for example
1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
are even some 2ppm/K parts but these seem unobtainium.
Are there any better datasheets? For example how does the tempco itself
vary with temperature? The company does not seem to have a web presence.
There
Regards,
John
WB6BNQ wb6bnq@cox.net writes:
Warren,
You need to do your schematics a little bit better. Besides, I
think you attached the wrong one. There are a number of errors on
the attached schematic.
Bill....WB6BNQ
WarrenS wrote:
Want a stable 10V ref? Best thing to do is get a fluke 732A/B and
have it calibrated regularly for a few years. BUT For the Extreme
(or cheap) Volt Nut, There is another choice. Attached is the
circuit of a 0.1 PPM, 10 volt transfer standard.
It has ONLY ONE critical part that limits its accuracy and
stability. Everything else can be made to have so little effect
as to be insignificant or else can be trimmed and calibrated out
without using a precision voltage source using a method similar to
what the daily auto cal in the HP3458 does. It measures it's own
internal 6+V reference.
Every year or more, best to record the value of the nominal 6.2
volts, to measurement the ageing rate, by using a known external
10 volt standard. To do that, an accurate external way is needed
to find the exact ratio of the 6.2 volts to the external 10 volt
standard. This can be done by using a high resolution Linear DMM
(aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null
meter (aka Fluke 845). Lets assume if one wants to do 0.1 PPM
things, then they have the equipment and skill needed to measure a
10v to 6v ratio to 0.1 PPM. After that ratio is known and
recorded the units internal 10 volts can then be set to that same
ratio without further need of the external 10V standard.
The main advantage of all this extra trouble is low cost accuracy
and stability. There are no critical or expensive resistor needed
or anything else critical, because all the TC and long term drift
are canceled out whenever a manual trim is done using an external
ratio device to check its own 10 V against it's own 6.2 volts.
Nothing very special is needed to build this beside care and time.
To get the best performance, It does take a lot of Nut-time to
match and select things. Total parts cost with a well supplied
analog junk box (or EBay) can be under $10.
Any good 6V type reference device could be used. I'm using a well
aged and selected 1n825 because I have a lot of history with them
and they work good without an oven. I have not done any mailing
test on these parts but they can be Zero TCed so that they do not
change 0.1PPM over normal room temperature and they show no change
when powered down for short periods of time and tend to return to
their original value when hit with a heat gun and cold stray, so
it makes a good part to start experimenting with.
Hi, could you be more specific?
m k <m1k3k1@hotmail.com> writes:
> Hi, The schematic has the opamp+ connected to the wrong point and
> there are some other details wrong.
Do you mean the +input or the +supply? They both look OK to me but noone
is saying what the problem is!
> In another job there was an instrument that used an 1n829a run at
> about 7.5mA for a 10 V reference that used to hold to better than 1uV
> over lab temperatures. It was done using the Op amp to generate the
> 10V that directly drove the 7.5mA into the zener, select on test
> resistors. but for 0.1ppm you need to address opamp heating and
> unintended thermocouples everywhere.
This is essentially what he is doing as far as I can see. But the opamp
heating is addressed by using the 78L05 as a "power driver", so the
opamp generates no heat! The heat-generating part is taken off-chip (or
off-board perhaps) so it does not affect the input offset voltage.
I agree 0.1ppm seems optimistic, but then I haven't built it. In my
admittedly limited experience thermocouples are not such a problem if
you keep everything isothermal, i.e. avoid temperature
gradients. Stating the obvious perhaps.
Even if it were 1ppm instead 0.1ppm I think this is still an order of
magnitude better than the homebrew ideas discussed so far (if
achievable).
John
>
>> From: john@devereux.me.uk
>> To: volt-nuts@febo.com
>> Date: Sun, 4 Sep 2011 15:56:00 +0100
>> Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
>>
>>
>> Hi Bill, Warren,
>>
>> Whats the matter with it? The schematic is clear enough to me,
>> interesting and seems quite clever to this wannabe nut.
>>
>> If I understand correctly, the opamp adjusts the divided-down output via
>> the regulator so that it matches the reference. Everything is powered
>> from the 10V output, eliminating problems with variable input voltages
>> and opamp PSRR.
>>
>> I would be interested to know:
>>
>> - Any suggestions for the wirewound resistors? Vishay bulk metal foil
>> resistors are not cheap but say 5ppm wirewounds are not pennies
>> either, I don't know how many you would need before a sufficient
>> number could be matched.
>>
>> - I am unclear as to the precise function of the 160k and 40M
>> resistors. I suppose one of them shifts the output, and the other
>> compensates the zener so it still runs at the same minimum-tempco
>> current?
>>
>> The zeners do look interesting as references, I admit I have been
>> fixated on LTZ1000s.
>>
>> Farnell/Newark don't stock the 1N825 but they do stock the
>> 1N825A.
>>
>> <http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825>
>>
>> According to its "datasheet"
>> <http://www.farnell.com/datasheets/332215.pdf> there are also some
>> similar looking parts that may have have improved tempco, for example
>> 1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
>> are even some 2ppm/K parts but these seem unobtainium.
>>
>> Are there any better datasheets? For example how does the tempco itself
>> vary with temperature? The company does not seem to have a web presence.
>>
>> There
>>
>> Regards,
>>
>> John
>>
>>
>> WB6BNQ <wb6bnq@cox.net> writes:
>>
>> > Warren,
>> >
>> > You need to do your schematics a little bit better. Besides, I
>> > think you attached the wrong one. There are a number of errors on
>> > the attached schematic.
>> >
>> > Bill....WB6BNQ
>> >
>> > WarrenS wrote:
>> >
>> >> Want a stable 10V ref? Best thing to do is get a fluke 732A/B and
>> >> have it calibrated regularly for a few years. BUT For the Extreme
>> >> (or cheap) Volt Nut, There is another choice. Attached is the
>> >> circuit of a 0.1 PPM, 10 volt transfer standard.
>> >>
>> >> It has ONLY ONE critical part that limits its accuracy and
>> >> stability. Everything else can be made to have so little effect
>> >> as to be insignificant or else can be trimmed and calibrated out
>> >> without using a precision voltage source using a method similar to
>> >> what the daily auto cal in the HP3458 does. It measures it's own
>> >> internal 6+V reference.
>> >>
>> >> Every year or more, best to record the value of the nominal 6.2
>> >> volts, to measurement the ageing rate, by using a known external
>> >> 10 volt standard. To do that, an accurate external way is needed
>> >> to find the exact ratio of the 6.2 volts to the external 10 volt
>> >> standard. This can be done by using a high resolution Linear DMM
>> >> (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null
>> >> meter (aka Fluke 845). Lets assume if one wants to do 0.1 PPM
>> >> things, then they have the equipment and skill needed to measure a
>> >> 10v to 6v ratio to 0.1 PPM. After that ratio is known and
>> >> recorded the units internal 10 volts can then be set to that same
>> >> ratio without further need of the external 10V standard.
>> >>
>> >> The main advantage of all this extra trouble is low cost accuracy
>> >> and stability. There are no critical or expensive resistor needed
>> >> or anything else critical, because all the TC and long term drift
>> >> are canceled out whenever a manual trim is done using an external
>> >> ratio device to check its own 10 V against it's own 6.2 volts.
>> >>
>> >> Nothing very special is needed to build this beside care and time.
>> >> To get the best performance, It does take a lot of Nut-time to
>> >> match and select things. Total parts cost with a well supplied
>> >> analog junk box (or EBay) can be under $10.
>> >>
>> >> Any good 6V type reference device could be used. I'm using a well
>> >> aged and selected 1n825 because I have a lot of history with them
>> >> and they work good without an oven. I have not done any mailing
>> >> test on these parts but they can be Zero TCed so that they do not
>> >> change 0.1PPM over normal room temperature and they show no change
>> >> when powered down for short periods of time and tend to return to
>> >> their original value when hit with a heat gun and cold stray, so
>> >> it makes a good part to start experimenting with.
--
John Devereux
PK
Poul-Henning Kamp
Sun, Sep 4, 2011 6:50 PM
Even if it were 1ppm instead 0.1ppm I think this is still an order of
magnitude better than the homebrew ideas discussed so far (if
achievable).
The problem is, to be sure you need to build at least three :-)
If somebody were to do a group-buy/kit of something like this,
I would buy a handful...
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
In message <871uvwp2xa.fsf@devereux.me.uk>, John Devereux writes:
>Even if it were 1ppm instead 0.1ppm I think this is still an order of
>magnitude better than the homebrew ideas discussed so far (if
>achievable).
The problem is, to be sure you need to build at least three :-)
If somebody were to do a group-buy/kit of something like this,
I would buy a handful...
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
JD
John Devereux
Sun, Sep 4, 2011 7:44 PM
Even if it were 1ppm instead 0.1ppm I think this is still an order of
magnitude better than the homebrew ideas discussed so far (if
achievable).
The problem is, to be sure you need to build at least three :-)
If somebody were to do a group-buy/kit of something like this,
I would buy a handful...
Well I've already ordered some zeners :)
Need a lot more experience/tinkering though. But perhaps some
"gold-plated" parts would do instead.
I'm thinking something like Vishay VAR or Z201 series resistors instead
of all the wirewounds, about 10 or 15 UKP each in singles, would need 3
or 4 I think.
The OP07 is looking a bit long in the tooth, there are more modern parts
or one could consider a chopper amp.
But I do like the idea of seeing what can be done with low cost parts
too.
--
John Devereux
"Poul-Henning Kamp" <phk@phk.freebsd.dk> writes:
> In message <871uvwp2xa.fsf@devereux.me.uk>, John Devereux writes:
>
>>Even if it were 1ppm instead 0.1ppm I think this is still an order of
>>magnitude better than the homebrew ideas discussed so far (if
>>achievable).
>
> The problem is, to be sure you need to build at least three :-)
>
> If somebody were to do a group-buy/kit of something like this,
> I would buy a handful...
Well I've already ordered some zeners :)
Need a lot more experience/tinkering though. But perhaps some
"gold-plated" parts would do instead.
I'm thinking something like Vishay VAR or Z201 series resistors instead
of all the wirewounds, about 10 or 15 UKP each in singles, would need 3
or 4 I think.
The OP07 is looking a bit long in the tooth, there are more modern parts
or one could consider a chopper amp.
But I do like the idea of seeing what can be done with low cost parts
too.
--
John Devereux
MK
m k
Sun, Sep 4, 2011 8:29 PM
Hi,
When looking closer at it the problem was the initial confusion caused by the guard rings as drawn. It was hard to be certain what should be connected where. When I drew it out for myself it all made sense.
Keeping everything isothermal is always difficult. But one answer is always the medicinal parraffin.
Hi, The schematic has the opamp+ connected to the wrong point and
there are some other details wrong.
Do you mean the +input or the +supply? They both look OK to me but noone
is saying what the problem is!
In another job there was an instrument that used an 1n829a run at
about 7.5mA for a 10 V reference that used to hold to better than 1uV
over lab temperatures. It was done using the Op amp to generate the
10V that directly drove the 7.5mA into the zener, select on test
resistors. but for 0.1ppm you need to address opamp heating and
unintended thermocouples everywhere.
This is essentially what he is doing as far as I can see. But the opamp
heating is addressed by using the 78L05 as a "power driver", so the
opamp generates no heat! The heat-generating part is taken off-chip (or
off-board perhaps) so it does not affect the input offset voltage.
I agree 0.1ppm seems optimistic, but then I haven't built it. In my
admittedly limited experience thermocouples are not such a problem if
you keep everything isothermal, i.e. avoid temperature
gradients. Stating the obvious perhaps.
Even if it were 1ppm instead 0.1ppm I think this is still an order of
magnitude better than the homebrew ideas discussed so far (if
achievable).
John
From: john@devereux.me.uk
To: volt-nuts@febo.com
Date: Sun, 4 Sep 2011 15:56:00 +0100
Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
Hi Bill, Warren,
Whats the matter with it? The schematic is clear enough to me,
interesting and seems quite clever to this wannabe nut.
If I understand correctly, the opamp adjusts the divided-down output via
the regulator so that it matches the reference. Everything is powered
from the 10V output, eliminating problems with variable input voltages
and opamp PSRR.
I would be interested to know:
-
Any suggestions for the wirewound resistors? Vishay bulk metal foil
resistors are not cheap but say 5ppm wirewounds are not pennies
either, I don't know how many you would need before a sufficient
number could be matched.
-
I am unclear as to the precise function of the 160k and 40M
resistors. I suppose one of them shifts the output, and the other
compensates the zener so it still runs at the same minimum-tempco
current?
The zeners do look interesting as references, I admit I have been
fixated on LTZ1000s.
Farnell/Newark don't stock the 1N825 but they do stock the
1N825A.
http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825
According to its "datasheet"
http://www.farnell.com/datasheets/332215.pdf there are also some
similar looking parts that may have have improved tempco, for example
1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
are even some 2ppm/K parts but these seem unobtainium.
Are there any better datasheets? For example how does the tempco itself
vary with temperature? The company does not seem to have a web presence.
There
Regards,
John
WB6BNQ wb6bnq@cox.net writes:
Warren,
You need to do your schematics a little bit better. Besides, I
think you attached the wrong one. There are a number of errors on
the attached schematic.
Bill....WB6BNQ
WarrenS wrote:
Want a stable 10V ref? Best thing to do is get a fluke 732A/B and
have it calibrated regularly for a few years. BUT For the Extreme
(or cheap) Volt Nut, There is another choice. Attached is the
circuit of a 0.1 PPM, 10 volt transfer standard.
It has ONLY ONE critical part that limits its accuracy and
stability. Everything else can be made to have so little effect
as to be insignificant or else can be trimmed and calibrated out
without using a precision voltage source using a method similar to
what the daily auto cal in the HP3458 does. It measures it's own
internal 6+V reference.
Every year or more, best to record the value of the nominal 6.2
volts, to measurement the ageing rate, by using a known external
10 volt standard. To do that, an accurate external way is needed
to find the exact ratio of the 6.2 volts to the external 10 volt
standard. This can be done by using a high resolution Linear DMM
(aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null
meter (aka Fluke 845). Lets assume if one wants to do 0.1 PPM
things, then they have the equipment and skill needed to measure a
10v to 6v ratio to 0.1 PPM. After that ratio is known and
recorded the units internal 10 volts can then be set to that same
ratio without further need of the external 10V standard.
The main advantage of all this extra trouble is low cost accuracy
and stability. There are no critical or expensive resistor needed
or anything else critical, because all the TC and long term drift
are canceled out whenever a manual trim is done using an external
ratio device to check its own 10 V against it's own 6.2 volts.
Nothing very special is needed to build this beside care and time.
To get the best performance, It does take a lot of Nut-time to
match and select things. Total parts cost with a well supplied
analog junk box (or EBay) can be under $10.
Any good 6V type reference device could be used. I'm using a well
aged and selected 1n825 because I have a lot of history with them
and they work good without an oven. I have not done any mailing
test on these parts but they can be Zero TCed so that they do not
change 0.1PPM over normal room temperature and they show no change
when powered down for short periods of time and tend to return to
their original value when hit with a heat gun and cold stray, so
it makes a good part to start experimenting with.
Hi,
When looking closer at it the problem was the initial confusion caused by the guard rings as drawn. It was hard to be certain what should be connected where. When I drew it out for myself it all made sense.
Keeping everything isothermal is always difficult. But one answer is always the medicinal parraffin.
> From: john@devereux.me.uk
> To: volt-nuts@febo.com
> Date: Sun, 4 Sep 2011 19:39:45 +0100
> Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
>
>
> Hi, could you be more specific?
>
> m k <m1k3k1@hotmail.com> writes:
>
> > Hi, The schematic has the opamp+ connected to the wrong point and
> > there are some other details wrong.
>
> Do you mean the +input or the +supply? They both look OK to me but noone
> is saying what the problem is!
>
> > In another job there was an instrument that used an 1n829a run at
> > about 7.5mA for a 10 V reference that used to hold to better than 1uV
> > over lab temperatures. It was done using the Op amp to generate the
> > 10V that directly drove the 7.5mA into the zener, select on test
> > resistors. but for 0.1ppm you need to address opamp heating and
> > unintended thermocouples everywhere.
>
> This is essentially what he is doing as far as I can see. But the opamp
> heating is addressed by using the 78L05 as a "power driver", so the
> opamp generates no heat! The heat-generating part is taken off-chip (or
> off-board perhaps) so it does not affect the input offset voltage.
>
> I agree 0.1ppm seems optimistic, but then I haven't built it. In my
> admittedly limited experience thermocouples are not such a problem if
> you keep everything isothermal, i.e. avoid temperature
> gradients. Stating the obvious perhaps.
>
> Even if it were 1ppm instead 0.1ppm I think this is still an order of
> magnitude better than the homebrew ideas discussed so far (if
> achievable).
>
> John
>
> >
> >> From: john@devereux.me.uk
> >> To: volt-nuts@febo.com
> >> Date: Sun, 4 Sep 2011 15:56:00 +0100
> >> Subject: Re: [volt-nuts] Sub PPM 10V transfer standard
> >>
> >>
> >> Hi Bill, Warren,
> >>
> >> Whats the matter with it? The schematic is clear enough to me,
> >> interesting and seems quite clever to this wannabe nut.
> >>
> >> If I understand correctly, the opamp adjusts the divided-down output via
> >> the regulator so that it matches the reference. Everything is powered
> >> from the 10V output, eliminating problems with variable input voltages
> >> and opamp PSRR.
> >>
> >> I would be interested to know:
> >>
> >> - Any suggestions for the wirewound resistors? Vishay bulk metal foil
> >> resistors are not cheap but say 5ppm wirewounds are not pennies
> >> either, I don't know how many you would need before a sufficient
> >> number could be matched.
> >>
> >> - I am unclear as to the precise function of the 160k and 40M
> >> resistors. I suppose one of them shifts the output, and the other
> >> compensates the zener so it still runs at the same minimum-tempco
> >> current?
> >>
> >> The zeners do look interesting as references, I admit I have been
> >> fixated on LTZ1000s.
> >>
> >> Farnell/Newark don't stock the 1N825 but they do stock the
> >> 1N825A.
> >>
> >> <http://uk.farnell.com/american-power-devices/1n825a/diode-zener-6-5v-0-25w-do35/dp/1651077?Ntt=1N825>
> >>
> >> According to its "datasheet"
> >> <http://www.farnell.com/datasheets/332215.pdf> there are also some
> >> similar looking parts that may have have improved tempco, for example
> >> 1N829A at 5ppm/K vs 1N825 at 20ppm/K. These are also available. There
> >> are even some 2ppm/K parts but these seem unobtainium.
> >>
> >> Are there any better datasheets? For example how does the tempco itself
> >> vary with temperature? The company does not seem to have a web presence.
> >>
> >> There
> >>
> >> Regards,
> >>
> >> John
> >>
> >>
> >> WB6BNQ <wb6bnq@cox.net> writes:
> >>
> >> > Warren,
> >> >
> >> > You need to do your schematics a little bit better. Besides, I
> >> > think you attached the wrong one. There are a number of errors on
> >> > the attached schematic.
> >> >
> >> > Bill....WB6BNQ
> >> >
> >> > WarrenS wrote:
> >> >
> >> >> Want a stable 10V ref? Best thing to do is get a fluke 732A/B and
> >> >> have it calibrated regularly for a few years. BUT For the Extreme
> >> >> (or cheap) Volt Nut, There is another choice. Attached is the
> >> >> circuit of a 0.1 PPM, 10 volt transfer standard.
> >> >>
> >> >> It has ONLY ONE critical part that limits its accuracy and
> >> >> stability. Everything else can be made to have so little effect
> >> >> as to be insignificant or else can be trimmed and calibrated out
> >> >> without using a precision voltage source using a method similar to
> >> >> what the daily auto cal in the HP3458 does. It measures it's own
> >> >> internal 6+V reference.
> >> >>
> >> >> Every year or more, best to record the value of the nominal 6.2
> >> >> volts, to measurement the ageing rate, by using a known external
> >> >> 10 volt standard. To do that, an accurate external way is needed
> >> >> to find the exact ratio of the 6.2 volts to the external 10 volt
> >> >> standard. This can be done by using a high resolution Linear DMM
> >> >> (aka HP3458A) OR a Kelvin-Varley divider (aka fluke 720 ) and null
> >> >> meter (aka Fluke 845). Lets assume if one wants to do 0.1 PPM
> >> >> things, then they have the equipment and skill needed to measure a
> >> >> 10v to 6v ratio to 0.1 PPM. After that ratio is known and
> >> >> recorded the units internal 10 volts can then be set to that same
> >> >> ratio without further need of the external 10V standard.
> >> >>
> >> >> The main advantage of all this extra trouble is low cost accuracy
> >> >> and stability. There are no critical or expensive resistor needed
> >> >> or anything else critical, because all the TC and long term drift
> >> >> are canceled out whenever a manual trim is done using an external
> >> >> ratio device to check its own 10 V against it's own 6.2 volts.
> >> >>
> >> >> Nothing very special is needed to build this beside care and time.
> >> >> To get the best performance, It does take a lot of Nut-time to
> >> >> match and select things. Total parts cost with a well supplied
> >> >> analog junk box (or EBay) can be under $10.
> >> >>
> >> >> Any good 6V type reference device could be used. I'm using a well
> >> >> aged and selected 1n825 because I have a lot of history with them
> >> >> and they work good without an oven. I have not done any mailing
> >> >> test on these parts but they can be Zero TCed so that they do not
> >> >> change 0.1PPM over normal room temperature and they show no change
> >> >> when powered down for short periods of time and tend to return to
> >> >> their original value when hit with a heat gun and cold stray, so
> >> >> it makes a good part to start experimenting with.
>
> --
>
> John Devereux
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.