Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior - every 3
or 4 seconds (it's variable), the pattern expands a few degrees in one
direction, then goes back in the other direction. It's NOT a sudden jump,
rather it is very much like watching someone's lungs taking a breath then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
The CW12-TIM is not a GPSDO, the frequency output is not disciplined, it
is, like the PPS, derived from the division of the free running internal
clock, driving the divisor (someway: added/skipped cycles) to align it with
the GPS. To mitigate the "breathing" there should be a sort of sawtooth
correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior - every 3
or 4 seconds (it's variable), the pattern expands a few degrees in one
direction, then goes back in the other direction. It's NOT a sudden jump,
rather it is very much like watching someone's lungs taking a breath then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to continue
to provide a GPS disciplined output
frequency with the GPS aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not disciplined, it
is, like the PPS, derived from the division of the free running internal
clock, driving the divisor (someway: added/skipped cycles) to align it with
the GPS. To mitigate the "breathing" there should be a sort of sawtooth
correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior - every
3
or 4 seconds (it's variable), the pattern expands a few degrees in one
direction, then goes back in the other direction. It's NOT a sudden jump,
rather it is very much like watching someone's lungs taking a breath then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi Pete,
Can you provide a picture of your Lissajous pattern? The Lissajous reflects
the phase of two signals. You use the term "breathing". As long as the
pattern does not collapse, you remain within 1Hz of the "reference" signal.
Once the signal collapses to a thin line, unless there is momentum in a
particular direction, you cannot tell if the drift has stopped and moved in
the other direction. If you never reach the collapse, you are within 1
cycle (not Hz!). If your pattern is going from a circle to an ellipse, and
back and forth it simply means your signal is drifting back and forth around
the reference.
If you are rolling over once a seconds, you are within 1Hz.
If you are rolling over once every ten seconds, you are within 1/10Hz.
100sec, 0.01Hz, etc...
Bob
-----Original Message-----
From: Peter McCollum via time-nuts time-nuts@lists.febo.com
Sent: Tuesday, January 14, 2025 9:52 PM
To: Discussion of precise time and frequency measurement
time-nuts@lists.febo.com
Cc: Peter McCollum saipan1959@gmail.com
Subject: [time-nuts] GPSDO "breathing" behavior
Importance: High
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged over many
seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior - every 3
or 4 seconds (it's variable), the pattern expands a few degrees in one
direction, then goes back in the other direction. It's NOT a sudden jump,
rather it is very much like watching someone's lungs taking a breath then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an
email to time-nuts-leave@lists.febo.com
If it is the credit card sized board of the CW12, then it cannot be a
disciplined oscillator time-nuts vise.
Here what Ed Palmer wrote on Sat Jun 26 2010 on this list about the
frequency output of the CW12:
Another GPS board with a 10 MHz output is the Navsync CW-12 module
(price ~US$85-90). I measured the 1 PPS output and found a Standard
Deviation of < 5 ns with a range of < 30 ns. The 10 MHz output is kept
on frequency by occasionally adjusting the period of the 10 MHz output
by one cycle of the CPU clock (~8.3 ns = 120 MHz). How often this
happens depends on the exact clock frequency of the particular unit. On
mine it happens about 200 times per second.
On Wed, Jan 15, 2025 at 5:51 PM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to continue
to provide a GPS disciplined output
frequency with the GPS aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not disciplined, it
is, like the PPS, derived from the division of the free running internal
clock, driving the divisor (someway: added/skipped cycles) to align it
with
the GPS. To mitigate the "breathing" there should be a sort of sawtooth
correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior -
every
3
or 4 seconds (it's variable), the pattern expands a few degrees in one
direction, then goes back in the other direction. It's NOT a sudden
jump,
rather it is very much like watching someone's lungs taking a breath
then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
GPSDO = GPS Disciplined Oscilator
They have:
A GPS module of some sort
A control loop that also acts as a noise filter
An independent oscillator that the control loop manages.
With just the module you are missing 2 and 3.
GPS modules have provided “chopped up” output signals for at least two decades and closer to three.
What they provide is not the same thing as what a GPSDO delivers.
Bob
On Jan 15, 2025, at 9:19 AM, Peter McCollum via time-nuts time-nuts@lists.febo.com wrote:
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to continue
to provide a GPS disciplined output
frequency with the GPS aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not disciplined, it
is, like the PPS, derived from the division of the free running internal
clock, driving the divisor (someway: added/skipped cycles) to align it with
the GPS. To mitigate the "breathing" there should be a sort of sawtooth
correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior - every
3
or 4 seconds (it's variable), the pattern expands a few degrees in one
direction, then goes back in the other direction. It's NOT a sudden jump,
rather it is very much like watching someone's lungs taking a breath then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Thanks for the comments so far folks! I am definitely NOT questioning
anyone's expertise, but I remain somewhat confused about the assertion that
the CW12-TIM is "not a GPSDO". If not, then what is it? It obviously
"disciplines an oscillator" in some way (see data below).
To clarify the "breathing" behavior:
Attached is a pic of the Lissajous, comparing the Sulzer vs the CW12-TIM
outputs, both at 1 Mhz. I have adjusted the fine-tuning on the Sulzer such
that the ellipse gradually changes shape over a period of many minutes.
BUT, over a period of perhaps 3 or 4 seconds, the ellipse goes back and
forth between the two versions shown in the picture. As I mentioned before,
it is NOT a sudden change, but rather looks like a lung expanding and
contracting somewhat.
I interpret that to mean that there is a temporary change in freq,
lasting less than 0.5 seconds or so, then the freq does the same thing in
the opposite direction, and the ellipse is back to it's original
shape/size. In the time frame of (say) a minute or two, there is NO overall
change to the ellipse.
This behavior is only seen with the CW12-TIM output - if I compare the
Sulzer against something else that is "pretty stable" (such as an HP
5345A), I don't see the "breathing".
Another data point:
If I power up the CW12-TIM with its antenna shielded to limit its access to
GPS signals, the Lissajous "rotates" at a rate of perhaps 1.5-2 Hz (out of
1 Mhz). Then, as soon as it starts to acquire sats, the Lissajous stops
moving... The datasheet says that the freq output is reliable when it has a
2-D solution.
Also attached is two pages from the CW12-TIM docs.
Is that NOT describing a "true GPSDO" ?
I see that the doc says "full PCR MTIE performance", although I certainly
don't understand the implications of that statement...
Pete
On Wed, Jan 15, 2025 at 1:06 PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:
Hi
GPSDO = GPS Disciplined Oscilator
They have:
A GPS module of some sort
A control loop that also acts as a noise filter
An independent oscillator that the control loop manages.
With just the module you are missing 2 and 3.
GPS modules have provided “chopped up” output signals for at least two
decades and closer to three.
What they provide is not the same thing as what a GPSDO delivers.
Bob
On Jan 15, 2025, at 9:19 AM, Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to continue
to provide a GPS disciplined output
frequency with the GPS aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not disciplined, it
is, like the PPS, derived from the division of the free running internal
clock, driving the divisor (someway: added/skipped cycles) to align it
with
the GPS. To mitigate the "breathing" there should be a sort of sawtooth
correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior -
every
3
or 4 seconds (it's variable), the pattern expands a few degrees in one
direction, then goes back in the other direction. It's NOT a sudden
jump,
rather it is very much like watching someone's lungs taking a breath
then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
What the modules do both on the PPS and on their other signals is to “dither” an internal oscillator.
The oscillator is not tuned to frequency. It free runs at some independent frequency (and phase).
The edges of the output are shifted from time to time to drop or add one period of the internal
oscillator.
This is not the same thing as tuning the oscillator on frequency. That’s why the “disciplining” term
is important in “GPSDO”.
Bob
On Jan 15, 2025, at 4:06 PM, Peter McCollum via time-nuts time-nuts@lists.febo.com wrote:
Thanks for the comments so far folks! I am definitely NOT questioning
anyone's expertise, but I remain somewhat confused about the assertion that
the CW12-TIM is "not a GPSDO". If not, then what is it? It obviously
"disciplines an oscillator" in some way (see data below).
To clarify the "breathing" behavior:
Attached is a pic of the Lissajous, comparing the Sulzer vs the CW12-TIM
outputs, both at 1 Mhz. I have adjusted the fine-tuning on the Sulzer such
that the ellipse gradually changes shape over a period of many minutes.
BUT, over a period of perhaps 3 or 4 seconds, the ellipse goes back and
forth between the two versions shown in the picture. As I mentioned before,
it is NOT a sudden change, but rather looks like a lung expanding and
contracting somewhat.
I interpret that to mean that there is a temporary change in freq,
lasting less than 0.5 seconds or so, then the freq does the same thing in
the opposite direction, and the ellipse is back to it's original
shape/size. In the time frame of (say) a minute or two, there is NO overall
change to the ellipse.
This behavior is only seen with the CW12-TIM output - if I compare the
Sulzer against something else that is "pretty stable" (such as an HP
5345A), I don't see the "breathing".
Another data point:
If I power up the CW12-TIM with its antenna shielded to limit its access to
GPS signals, the Lissajous "rotates" at a rate of perhaps 1.5-2 Hz (out of
1 Mhz). Then, as soon as it starts to acquire sats, the Lissajous stops
moving... The datasheet says that the freq output is reliable when it has a
2-D solution.
Also attached is two pages from the CW12-TIM docs.
Is that NOT describing a "true GPSDO" ?
I see that the doc says "full PCR MTIE performance", although I certainly
don't understand the implications of that statement...
Pete
On Wed, Jan 15, 2025 at 1:06 PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:
Hi
GPSDO = GPS Disciplined Oscilator
They have:
A GPS module of some sort
A control loop that also acts as a noise filter
An independent oscillator that the control loop manages.
With just the module you are missing 2 and 3.
GPS modules have provided “chopped up” output signals for at least two
decades and closer to three.
What they provide is not the same thing as what a GPSDO delivers.
Bob
On Jan 15, 2025, at 9:19 AM, Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to continue
to provide a GPS disciplined output
frequency with the GPS aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not disciplined, it
is, like the PPS, derived from the division of the free running internal
clock, driving the divisor (someway: added/skipped cycles) to align it
with
the GPS. To mitigate the "breathing" there should be a sort of sawtooth
correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior -
every
3
or 4 seconds (it's variable), the pattern expands a few degrees in one
direction, then goes back in the other direction. It's NOT a sudden
jump,
rather it is very much like watching someone's lungs taking a breath
then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
<Lissajous.JPG><p2.jpg><p1.jpg>_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
I am probably one of the least qualified members on this list to comment.
HOWEVER, I think that the the fact that the CW12-TIM is meant to be a
((an) almost, maybe exact or better) drop in replacement for a Motorola
M12+, which by design has NEVER been or ASSERTED to be ANYTHING other than
only the GPS module for a GPSDO or frequency standard should be a MAJOR
clue. Just my $.02
Thomas Valerio
Thanks for the comments so far folks! I am definitely NOT questioning
anyone's expertise, but I remain somewhat confused about the assertion
that
the CW12-TIM is "not a GPSDO". If not, then what is it? It obviously
"disciplines an oscillator" in some way (see data below).
To clarify the "breathing" behavior:
Attached is a pic of the Lissajous, comparing the Sulzer vs the CW12-TIM
outputs, both at 1 Mhz. I have adjusted the fine-tuning on the Sulzer such
that the ellipse gradually changes shape over a period of many minutes.
BUT, over a period of perhaps 3 or 4 seconds, the ellipse goes back and
forth between the two versions shown in the picture. As I mentioned
before,
it is NOT a sudden change, but rather looks like a lung expanding and
contracting somewhat.
I interpret that to mean that there is a temporary change in freq,
lasting less than 0.5 seconds or so, then the freq does the same thing in
the opposite direction, and the ellipse is back to it's original
shape/size. In the time frame of (say) a minute or two, there is NO
overall
change to the ellipse.
This behavior is only seen with the CW12-TIM output - if I compare the
Sulzer against something else that is "pretty stable" (such as an HP
5345A), I don't see the "breathing".
Another data point:
If I power up the CW12-TIM with its antenna shielded to limit its access
to
GPS signals, the Lissajous "rotates" at a rate of perhaps 1.5-2 Hz (out of
1 Mhz). Then, as soon as it starts to acquire sats, the Lissajous stops
moving... The datasheet says that the freq output is reliable when it has
a
2-D solution.
Also attached is two pages from the CW12-TIM docs.
Is that NOT describing a "true GPSDO" ?
I see that the doc says "full PCR MTIE performance", although I certainly
don't understand the implications of that statement...
Pete
On Wed, Jan 15, 2025 at 1:06â¯PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:
Hi
GPSDO = GPS Disciplined Oscilator
They have:
A GPS module of some sort
A control loop that also acts as a noise filter
An independent oscillator that the control loop manages.
With just the module you are missing 2 and 3.
GPS modules have provided âchopped upâ output signals for at least
two
decades and closer to three.
What they provide is not the same thing as what a GPSDO delivers.
Bob
On Jan 15, 2025, at 9:19â¯AM, Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to
continue
to provide a GPS disciplined output
frequency with the GPS aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01â¯AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not disciplined,
it
is, like the PPS, derived from the division of the free running
internal
clock, driving the divisor (someway: added/skipped cycles) to align
it
with
the GPS. To mitigate the "breathing" there should be a sort of
sawtooth
correction also for the frequency output (not possible for that
speed).
On Wed, Jan 15, 2025 at 8:32â¯AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged
over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior -
every
3
or 4 seconds (it's variable), the pattern expands a few degrees in
one
direction, then goes back in the other direction. It's NOT a sudden
jump,
rather it is very much like watching someone's lungs taking a breath
then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi Bob,
I agree with you, you don't "tune" the oscillator. To expand, the NCO has two parts, the oscillator, and the numerical control. The NCO is a very long digital divider, with single or multiple loops. The oscillator can be at any high frequency, and the NCO "tuning" is what gives you the span of 10Hz to 10MHz.
The oscillator will drift, but the processor will adjust the output by "tuning" the NCO. The tuning is simply writing new digital values to the NCO. I assume, whether the oscillator is built in to the NCO or separate, it's a tiny little oscillator, probably not even a TCXO. So, it drifts, again, I assume for the most part, in one direction. The processor will then periodically (1PPS) adjust the NCO before the phase rolls over. The oscillator, NCO, and the processor form what looks very much like a frequency synthesizer.
You mentioned "disciplining". Isn't that more of a processor disciplining? The processor learns the drift, and when in hold over, continues to adjust the EFC to zero the anticipated drift? Is the prediction function active when the 1PPS is available, or only when in holdover? Maybe it's all semantics!
Bob B
-----Original Message-----
From: Bob Camp via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, January 15, 2025 5:41 PM
To: Discussion of precise time and frequency measurement time-nuts@lists.febo.com
Cc: Bob Camp kb8tq@n1k.org
Subject: [time-nuts] Re: GPSDO "breathing" behavior
Importance: High
Hi
What the modules do both on the PPS and on their other signals is to “dither” an internal oscillator.
The oscillator is not tuned to frequency. It free runs at some independent frequency (and phase).
The edges of the output are shifted from time to time to drop or add one period of the internal oscillator.
This is not the same thing as tuning the oscillator on frequency. That’s why the “disciplining” term is important in “GPSDO”.
Bob
On Jan 15, 2025, at 4:06 PM, Peter McCollum via time-nuts time-nuts@lists.febo.com wrote:
Thanks for the comments so far folks! I am definitely NOT questioning
anyone's expertise, but I remain somewhat confused about the assertion
that the CW12-TIM is "not a GPSDO". If not, then what is it? It
obviously "disciplines an oscillator" in some way (see data below).
To clarify the "breathing" behavior:
Attached is a pic of the Lissajous, comparing the Sulzer vs the
CW12-TIM outputs, both at 1 Mhz. I have adjusted the fine-tuning on
the Sulzer such that the ellipse gradually changes shape over a period of many minutes.
BUT, over a period of perhaps 3 or 4 seconds, the ellipse goes back
and forth between the two versions shown in the picture. As I
mentioned before, it is NOT a sudden change, but rather looks like a
lung expanding and contracting somewhat.
I interpret that to mean that there is a temporary change in freq,
lasting less than 0.5 seconds or so, then the freq does the same thing
in the opposite direction, and the ellipse is back to it's original
shape/size. In the time frame of (say) a minute or two, there is NO
overall change to the ellipse.
This behavior is only seen with the CW12-TIM output - if I compare the
Sulzer against something else that is "pretty stable" (such as an HP
5345A), I don't see the "breathing".
Another data point:
If I power up the CW12-TIM with its antenna shielded to limit its
access to GPS signals, the Lissajous "rotates" at a rate of perhaps
1.5-2 Hz (out of
1 Mhz). Then, as soon as it starts to acquire sats, the Lissajous
stops moving... The datasheet says that the freq output is reliable
when it has a 2-D solution.
Also attached is two pages from the CW12-TIM docs.
Is that NOT describing a "true GPSDO" ?
I see that the doc says "full PCR MTIE performance", although I
certainly don't understand the implications of that statement...
Pete
On Wed, Jan 15, 2025 at 1:06 PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:
Hi
GPSDO = GPS Disciplined Oscilator
They have:
A GPS module of some sort
A control loop that also acts as a noise filter
An independent oscillator that the control loop manages.
With just the module you are missing 2 and 3.
GPS modules have provided “chopped up” output signals for at least
two decades and closer to three.
What they provide is not the same thing as what a GPSDO delivers.
Bob
On Jan 15, 2025, at 9:19 AM, Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to
continue to provide a GPS disciplined output frequency with the GPS
aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not
disciplined, it is, like the PPS, derived from the division of the
free running internal clock, driving the divisor (someway:
added/skipped cycles) to align it
with
the GPS. To mitigate the "breathing" there should be a sort of
sawtooth correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs
a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged
over many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior -
every
3
or 4 seconds (it's variable), the pattern expands a few degrees in
one direction, then goes back in the other direction. It's NOT a
sudden
jump,
rather it is very much like watching someone's lungs taking a
breath
then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO
is repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
<Lissajous.JPG><p2.jpg><p1.jpg>_______________________________________
________ time-nuts mailing list -- time-nuts@lists.febo.com To
unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an email to time-nuts-leave@lists.febo.com
Another way to look at this is to say a real VCXO has an analog input, subject to noise, pushing and pullin, potentially finite DAC resolution, and a not insignificant Q associated with the crystal.
If you were to average a crystal oscillator frequency and sampled say every 100 seconds, you’d find the tuning steps approximate a nearly continuous adjust range. A digital NCO at best is one count ahead or one count behind. The shorter the time interval the worse it looks. If you divided a higher frequency NCO to obtain a lower frequency output, you’d improve by the division ratio given fixed sample interval.
Good frequency planning can probably help in rendering a system amenable to analysis, but ultimately even with good clock planning you're going to have to deal with loop bandwidth and that pesky +-1 count per interval, and it's transformation from time domain to frequency domain.
From: rbenward--- via time-nuts time-nuts@lists.febo.com
Sent: Thursday, January 16, 2025 10:56:04 AM
To: 'Discussion of precise time and frequency measurement' time-nuts@lists.febo.com
Cc: rbenward@verizon.net rbenward@verizon.net
Subject: [time-nuts] Re: GPSDO "breathing" behavior
Hi Bob,
I agree with you, you don't "tune" the oscillator. To expand, the NCO has two parts, the oscillator, and the numerical control. The NCO is a very long digital divider, with single or multiple loops. The oscillator can be at any high frequency, and the NCO "tuning" is what gives you the span of 10Hz to 10MHz.
The oscillator will drift, but the processor will adjust the output by "tuning" the NCO. The tuning is simply writing new digital values to the NCO. I assume, whether the oscillator is built in to the NCO or separate, it's a tiny little oscillator, probably not even a TCXO. So, it drifts, again, I assume for the most part, in one direction. The processor will then periodically (1PPS) adjust the NCO before the phase rolls over. The oscillator, NCO, and the processor form what looks very much like a frequency synthesizer.
You mentioned "disciplining". Isn't that more of a processor disciplining? The processor learns the drift, and when in hold over, continues to adjust the EFC to zero the anticipated drift? Is the prediction function active when the 1PPS is available, or only when in holdover? Maybe it's all semantics!
Bob B
-----Original Message-----
From: Bob Camp via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, January 15, 2025 5:41 PM
To: Discussion of precise time and frequency measurement time-nuts@lists.febo.com
Cc: Bob Camp kb8tq@n1k.org
Subject: [time-nuts] Re: GPSDO "breathing" behavior
Importance: High
Hi
What the modules do both on the PPS and on their other signals is to “dither” an internal oscillator.
The oscillator is not tuned to frequency. It free runs at some independent frequency (and phase).
The edges of the output are shifted from time to time to drop or add one period of the internal oscillator.
This is not the same thing as tuning the oscillator on frequency. That’s why the “disciplining” term is important in “GPSDO”.
Bob
On Jan 15, 2025, at 4:06 PM, Peter McCollum via time-nuts time-nuts@lists.febo.com wrote:
Thanks for the comments so far folks! I am definitely NOT questioning
anyone's expertise, but I remain somewhat confused about the assertion
that the CW12-TIM is "not a GPSDO". If not, then what is it? It
obviously "disciplines an oscillator" in some way (see data below).
To clarify the "breathing" behavior:
Attached is a pic of the Lissajous, comparing the Sulzer vs the
CW12-TIM outputs, both at 1 Mhz. I have adjusted the fine-tuning on
the Sulzer such that the ellipse gradually changes shape over a period of many minutes.
BUT, over a period of perhaps 3 or 4 seconds, the ellipse goes back
and forth between the two versions shown in the picture. As I
mentioned before, it is NOT a sudden change, but rather looks like a
lung expanding and contracting somewhat.
I interpret that to mean that there is a temporary change in freq,
lasting less than 0.5 seconds or so, then the freq does the same thing
in the opposite direction, and the ellipse is back to it's original
shape/size. In the time frame of (say) a minute or two, there is NO
overall change to the ellipse.
This behavior is only seen with the CW12-TIM output - if I compare the
Sulzer against something else that is "pretty stable" (such as an HP
5345A), I don't see the "breathing".
Another data point:
If I power up the CW12-TIM with its antenna shielded to limit its
access to GPS signals, the Lissajous "rotates" at a rate of perhaps
1.5-2 Hz (out of
1 Mhz). Then, as soon as it starts to acquire sats, the Lissajous
stops moving... The datasheet says that the freq output is reliable
when it has a 2-D solution.
Also attached is two pages from the CW12-TIM docs.
Is that NOT describing a "true GPSDO" ?
I see that the doc says "full PCR MTIE performance", although I
certainly don't understand the implications of that statement...
Pete
On Wed, Jan 15, 2025 at 1:06 PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:
Hi
GPSDO = GPS Disciplined Oscilator
They have:
A GPS module of some sort
A control loop that also acts as a noise filter
An independent oscillator that the control loop manages.
With just the module you are missing 2 and 3.
GPS modules have provided “chopped up” output signals for at least
two decades and closer to three.
What they provide is not the same thing as what a GPSDO delivers.
Bob
On Jan 15, 2025, at 9:19 AM, Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to
continue to provide a GPS disciplined output frequency with the GPS
aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not
disciplined, it is, like the PPS, derived from the division of the
free running internal clock, driving the divisor (someway:
added/skipped cycles) to align it
with
the GPS. To mitigate the "breathing" there should be a sort of
sawtooth correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs
a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged
over many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior -
every
3
or 4 seconds (it's variable), the pattern expands a few degrees in
one direction, then goes back in the other direction. It's NOT a
sudden
jump,
rather it is very much like watching someone's lungs taking a
breath
then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO
is repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
https://urldefense.proofpoint.com/v2/url?u=http-3A__Lissajous.JPG&d=DwIGaQ&c=euGZstcaTDllvimEN8b7jXrwqOf-v5A_CdpgnVfiiMM&r=JsDsKeR7cZC8wbZhIlxxBQ&m=yVh8m8tmSsEJ7hVfAOlEU8Yu0BRbEo6imagx9T1UmSjyXgVofche0UTjoChwhtPC&s=XIJgAKCNHYiFSQbTueQahZaV1baoT4TDR-5QIb2uv7k&e=https://urldefense.proofpoint.com/v2/url?u=http-3A__p2.jpg&d=DwIGaQ&c=euGZstcaTDllvimEN8b7jXrwqOf-v5A_CdpgnVfiiMM&r=JsDsKeR7cZC8wbZhIlxxBQ&m=yVh8m8tmSsEJ7hVfAOlEU8Yu0BRbEo6imagx9T1UmSjyXgVofche0UTjoChwhtPC&s=2tKTcEL1FZSJkIl5Z1GuTTPRoo3NGvUrn1M78-FpW7Q&e=https://urldefense.proofpoint.com/v2/url?u=http-3A__p1.jpg&d=DwIGaQ&c=euGZstcaTDllvimEN8b7jXrwqOf-v5A_CdpgnVfiiMM&r=JsDsKeR7cZC8wbZhIlxxBQ&m=yVh8m8tmSsEJ7hVfAOlEU8Yu0BRbEo6imagx9T1UmSjyXgVofche0UTjoChwhtPC&s=CWqDfQKJ3oIz8lF6ks5e50WakcdqoEn8WtAJdyCiEQI&e=_______________________________________
________ time-nuts mailing list -- time-nuts@lists.febo.com To
unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
Hi
The purpose of the control loop in a GPSDO is to eliminate noise, not to add it. A
GPSDO has a clean output spectrum rather than a total mess. That’s part of why
it is what it is.
The processing in the GPS module does no noise reduction at all. You get all the
trash the GPS puts out based on it’s second to second estimates.
Bob
On Jan 16, 2025, at 11:56 AM, rbenward--- via time-nuts time-nuts@lists.febo.com wrote:
Hi Bob,
I agree with you, you don't "tune" the oscillator. To expand, the NCO has two parts, the oscillator, and the numerical control. The NCO is a very long digital divider, with single or multiple loops. The oscillator can be at any high frequency, and the NCO "tuning" is what gives you the span of 10Hz to 10MHz.
The oscillator will drift, but the processor will adjust the output by "tuning" the NCO. The tuning is simply writing new digital values to the NCO. I assume, whether the oscillator is built in to the NCO or separate, it's a tiny little oscillator, probably not even a TCXO. So, it drifts, again, I assume for the most part, in one direction. The processor will then periodically (1PPS) adjust the NCO before the phase rolls over. The oscillator, NCO, and the processor form what looks very much like a frequency synthesizer.
You mentioned "disciplining". Isn't that more of a processor disciplining? The processor learns the drift, and when in hold over, continues to adjust the EFC to zero the anticipated drift? Is the prediction function active when the 1PPS is available, or only when in holdover? Maybe it's all semantics!
Bob B
-----Original Message-----
From: Bob Camp via time-nuts time-nuts@lists.febo.com
Sent: Wednesday, January 15, 2025 5:41 PM
To: Discussion of precise time and frequency measurement time-nuts@lists.febo.com
Cc: Bob Camp kb8tq@n1k.org
Subject: [time-nuts] Re: GPSDO "breathing" behavior
Importance: High
Hi
What the modules do both on the PPS and on their other signals is to “dither” an internal oscillator.
The oscillator is not tuned to frequency. It free runs at some independent frequency (and phase).
The edges of the output are shifted from time to time to drop or add one period of the internal oscillator.
This is not the same thing as tuning the oscillator on frequency. That’s why the “disciplining” term is important in “GPSDO”.
Bob
On Jan 15, 2025, at 4:06 PM, Peter McCollum via time-nuts time-nuts@lists.febo.com wrote:
Thanks for the comments so far folks! I am definitely NOT questioning
anyone's expertise, but I remain somewhat confused about the assertion
that the CW12-TIM is "not a GPSDO". If not, then what is it? It
obviously "disciplines an oscillator" in some way (see data below).
To clarify the "breathing" behavior:
Attached is a pic of the Lissajous, comparing the Sulzer vs the
CW12-TIM outputs, both at 1 Mhz. I have adjusted the fine-tuning on
the Sulzer such that the ellipse gradually changes shape over a period of many minutes.
BUT, over a period of perhaps 3 or 4 seconds, the ellipse goes back
and forth between the two versions shown in the picture. As I
mentioned before, it is NOT a sudden change, but rather looks like a
lung expanding and contracting somewhat.
I interpret that to mean that there is a temporary change in freq,
lasting less than 0.5 seconds or so, then the freq does the same thing
in the opposite direction, and the ellipse is back to it's original
shape/size. In the time frame of (say) a minute or two, there is NO
overall change to the ellipse.
This behavior is only seen with the CW12-TIM output - if I compare the
Sulzer against something else that is "pretty stable" (such as an HP
5345A), I don't see the "breathing".
Another data point:
If I power up the CW12-TIM with its antenna shielded to limit its
access to GPS signals, the Lissajous "rotates" at a rate of perhaps
1.5-2 Hz (out of
1 Mhz). Then, as soon as it starts to acquire sats, the Lissajous
stops moving... The datasheet says that the freq output is reliable
when it has a 2-D solution.
Also attached is two pages from the CW12-TIM docs.
Is that NOT describing a "true GPSDO" ?
I see that the doc says "full PCR MTIE performance", although I
certainly don't understand the implications of that statement...
Pete
On Wed, Jan 15, 2025 at 1:06 PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:
Hi
GPSDO = GPS Disciplined Oscilator
They have:
A GPS module of some sort
A control loop that also acts as a noise filter
An independent oscillator that the control loop manages.
With just the module you are missing 2 and 3.
GPS modules have provided “chopped up” output signals for at least
two decades and closer to three.
What they provide is not the same thing as what a GPSDO delivers.
Bob
On Jan 15, 2025, at 9:19 AM, Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to
continue to provide a GPS disciplined output frequency with the GPS
aerial located completely inside a building..."
Is it just a matter of semantics, or is there a firm definition of
"disciplined"?
Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:
The CW12-TIM is not a GPSDO, the frequency output is not
disciplined, it is, like the PPS, derived from the division of the
free running internal clock, driving the divisor (someway:
added/skipped cycles) to align it
with
the GPS. To mitigate the "breathing" there should be a sort of
sawtooth correction also for the frequency output (not possible for that speed).
On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:
Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs
a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged
over many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior -
every
3
or 4 seconds (it's variable), the pattern expands a few degrees in
one direction, then goes back in the other direction. It's NOT a
sudden
jump,
rather it is very much like watching someone's lungs taking a
breath
then
letting it out.
Is it normal, or does it imply that the "discipline" in the GPSDO
is repeatedly needing a correction?
Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe
send an email to time-nuts-leave@lists.febo.com
<Lissajous.JPG><p2.jpg><p1.jpg>_______________________________________
________ time-nuts mailing list -- time-nuts@lists.febo.com To
unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
On 2025-01-15 22:38, Thomas Valerio via time-nuts wrote:
I am probably one of the least qualified members on this list to
comment.
HOWEVER, I think that the the fact that the CW12-TIM is meant to be a
((an) almost, maybe exact or better) drop in replacement for a Motorola
M12+, which by design has NEVER been or ASSERTED to be ANYTHING other
than
only the GPS module for a GPSDO or frequency standard should be a MAJOR
clue. Just my $.02
This explains and demonstrates the M12 sawtooth pretty well.
http://www.leapsecond.com/pages/m12/sawtooth.htm
Marshall Eubanks
Thomas Valerio
Thanks for the comments so far folks! I am definitely NOT questioning
anyone's expertise, but I remain somewhat confused about the assertion
that
the CW12-TIM is "not a GPSDO". If not, then what is it? It obviously
"disciplines an oscillator" in some way (see data below).To clarify the "breathing" behavior:
Attached is a pic of the Lissajous, comparing the Sulzer vs the
CW12-TIM
outputs, both at 1 Mhz. I have adjusted the fine-tuning on the Sulzer
such
that the ellipse gradually changes shape over a period of many
minutes.
BUT, over a period of perhaps 3 or 4 seconds, the ellipse goes back
and
forth between the two versions shown in the picture. As I mentioned
before,
it is NOT a sudden change, but rather looks like a lung expanding and
contracting somewhat.
I interpret that to mean that there is a temporary change in freq,
lasting less than 0.5 seconds or so, then the freq does the same thing
in
the opposite direction, and the ellipse is back to it's original
shape/size. In the time frame of (say) a minute or two, there is NO
overall
change to the ellipse.
This behavior is only seen with the CW12-TIM output - if I compare the
Sulzer against something else that is "pretty stable" (such as an HP
5345A), I don't see the "breathing".Another data point:
If I power up the CW12-TIM with its antenna shielded to limit its
access
to
GPS signals, the Lissajous "rotates" at a rate of perhaps 1.5-2 Hz
(out of
1 Mhz). Then, as soon as it starts to acquire sats, the Lissajous
stops
moving... The datasheet says that the freq output is reliable when it
has
a
2-D solution.Also attached is two pages from the CW12-TIM docs.
Is that NOT describing a "true GPSDO" ?
I see that the doc says "full PCR MTIE performance", although I
certainly
don't understand the implications of that statement...Pete
On Wed, Jan 15, 2025 at 1:06 PM Bob Camp via time-nuts <
time-nuts@lists.febo.com> wrote:Hi
GPSDO = GPS Disciplined Oscilator
They have:
A GPS module of some sort
A control loop that also acts as a noise filter
An independent oscillator that the control loop manages.
With just the module you are missing 2 and 3.
GPS modules have provided “chopped up†output signals for at
least
two
decades and closer to three.
What they provide is not the same thing as what a GPSDO delivers.Bob
On Jan 15, 2025, at 9:19 AM, Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:Azelio wrote:
The CW12-TIM is not a GPSDO
From the CW12-TIM manual:
"The graph below demonstrates the ability of the CW12-TIM GPS to
continue
to provide a GPS disciplined output
frequency with the GPS aerial located completely inside a building..."Is it just a matter of semantics, or is there a firm definition of
"disciplined"?Pete
On Wed, Jan 15, 2025 at 3:01 AM Azelio Boriani via time-nuts <
time-nuts@lists.febo.com> wrote:The CW12-TIM is not a GPSDO, the frequency output is not disciplined,
it
is, like the PPS, derived from the division of the free running
internal
clock, driving the divisor (someway: added/skipped cycles) to align
it
with
the GPS. To mitigate the "breathing" there should be a sort of
sawtooth
correction also for the frequency output (not possible for that
speed).On Wed, Jan 15, 2025 at 8:32 AM Peter McCollum via time-nuts <
time-nuts@lists.febo.com> wrote:Just curious if this is normal behavior.
I use an X-Y display to show a Lissajous of my CW12-TIM module vs a
fairly
stable oscillator (Sulzer 5A, or Spectracom 8161), all at 1 MHz.
The CW12's output gives an (apparently) very solid output averaged
over
many seconds or longer.
But in the short term, the Lissajous does a "breathing" behavior -
every
3
or 4 seconds (it's variable), the pattern expands a few degrees in
one
direction, then goes back in the other direction. It's NOT a sudden
jump,
rather it is very much like watching someone's lungs taking a breath
then
letting it out.Is it normal, or does it imply that the "discipline" in the GPSDO is
repeatedly needing a correction?Thanks for any comments!
Pete
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com