DIY Low offset Phase Noise Analyzer (Erik Kaashoek)

BK
Bob kb8tq
Sun, Jul 10, 2022 4:07 PM

Hi

Yes it is a pain to implement dual supplies. I ponder that issue every time
I build one of these setups. I’ve built a lot of them …. If you are going to
do a single supply, setting up a “virtual ground” is probably the best way
to go. Do it with a drive circuit to provide very clean 15V off of a 30V supply
then tack everything ( including all the mixer grounds to that 15V supply.

Keeping the signal undistorted before you check the beat note and use it
to drive the EFC does keep you out of various issues. You do not want to
deal with possible clipping / saturation artifacts getting into either process.

With devices having positive side EFC, negative side EFC, and “both
sides” EFC, it’s hard to get around a dual supply in any sort of general
purpose device. A 15V center point is not going to fit any EFC that I’ve
seen :).

Struggling with the ground loop problem is always the big deal in any
setup. Trying to rule out / take out line noise is usually the final straw in
any series of tests. Doing that with everything at “real ground” is just a
bit easier.

Part of the calibration is measuring the beat note as it goes past zero.
The ’scope gets cranked up and you look at a bit of the crossing right
at ground. Keeping the device happy while doing this is much easier if
the chassis does not need to float at 15V.

Whatever is used as a supply turns out to be a dedicated device. The
same ground loop / isolation stuff get in here. An old style non-switching
design is just about mandatory. Keeping switching artifacts out of things
is almost impossible. All of this makes a “build from scratch” approach
less and less crazy. Old style three terminal regulators ( so 78x18 / 79x18 )
are not as easy to find these days. They do fine if you happen to have a
pair …. There’s really not much power used by any of this. The need for
anything massive. 100 ma out of each side is overkill ….

As you build things up, you eventually come to the realization that a big
sheet of brass is a good idea for the ground. Tie this and that to the sheet.
Keep everything non-essential away and likely keep it turned off. Tying
a dedicated supply to that sheet along with the amp and EFC stuff is not
at all unusual.

Bob

On Jul 9, 2022, at 11:11 PM, Erik Kaashoek erik@kaashoek.com wrote:

Hi Magnus,
Yes, and it works very well, locking is easier as once locked it nicely stay's in lock, , even with a slow drift of either the DUT or the reference. As I could not find a bipolar capacitor the tuning potmeter has to be kept at the low side to avoid blowing the integration capacitor. Maybe a back to back series capacitor with pull down resistor is safer to use.
Will need to update the schematic to show the small improvements.

@Bob,
You mentioned "dual supplies with high voltage" for the first gain opamp. How much impact would dual voltage bring as its a pain to implement.
I understand everything gets ground reference and you loose the noise of the buffer opamp but as the first gain opamp is in differential mode for its input it does not see the noise of the buffer opamp. Or am I making a mistake?

On 10-7-2022 2:02, Magnus Danielson via time-nuts wrote:

Have you attempted doing a PI-loop as I've suggested?

Hi Yes it is a pain to implement dual supplies. I ponder that issue every time I build one of these setups. I’ve built a lot of them …. If you are going to do a single supply, setting up a “virtual ground” is probably the best way to go. Do it with a drive circuit to provide very clean 15V off of a 30V supply then tack everything ( including *all* the mixer grounds to that 15V supply. Keeping the signal undistorted before you check the beat note and use it to drive the EFC does keep you out of various issues. You do not want to deal with possible clipping / saturation artifacts getting into either process. With devices having positive side EFC, negative side EFC, and “both sides” EFC, it’s hard to get around a dual supply in any sort of general purpose device. A 15V center point is not going to fit any EFC that I’ve seen :). Struggling with the ground loop problem is always the big deal in any setup. Trying to rule out / take out line noise is usually the final straw in any series of tests. Doing that with everything at “real ground” is just a bit easier. Part of the calibration is measuring the beat note as it goes past zero. The ’scope gets cranked up and you look at a bit of the crossing right at ground. Keeping the device happy while doing this is much easier if the chassis does not need to float at 15V. Whatever is used as a supply turns out to be a dedicated device. The same ground loop / isolation stuff get in here. An old style non-switching design is just about mandatory. Keeping switching artifacts out of things is almost impossible. All of this makes a “build from scratch” approach less and less crazy. Old style three terminal regulators ( so 78x18 / 79x18 ) are not as easy to find these days. They do fine if you happen to have a pair …. There’s really not much power used by any of this. The need for anything massive. 100 ma out of each side is overkill …. As you build things up, you eventually come to the realization that a big sheet of brass is a good idea for the ground. Tie this and that to the sheet. Keep everything non-essential away and likely keep it turned off. Tying a dedicated supply to that sheet along with the amp and EFC stuff is not at all unusual. Bob > On Jul 9, 2022, at 11:11 PM, Erik Kaashoek <erik@kaashoek.com> wrote: > > Hi Magnus, > Yes, and it works very well, locking is easier as once locked it nicely stay's in lock, , even with a slow drift of either the DUT or the reference. As I could not find a bipolar capacitor the tuning potmeter has to be kept at the low side to avoid blowing the integration capacitor. Maybe a back to back series capacitor with pull down resistor is safer to use. > Will need to update the schematic to show the small improvements. > > @Bob, > You mentioned "dual supplies with high voltage" for the first gain opamp. How much impact would dual voltage bring as its a pain to implement. > I understand everything gets ground reference and you loose the noise of the buffer opamp but as the first gain opamp is in differential mode for its input it does not see the noise of the buffer opamp. Or am I making a mistake? > > On 10-7-2022 2:02, Magnus Danielson via time-nuts wrote: >> Have you attempted doing a PI-loop as I've suggested? >
LJ
Lux, Jim
Sun, Jul 10, 2022 5:14 PM

On 7/10/22 9:07 AM, Bob kb8tq via time-nuts wrote:

Hi

Yes it is a pain to implement dual supplies. I ponder that issue every time
I build one of these setups. I’ve built a lot of them …. If you are going to
do a single supply, setting up a “virtual ground” is probably the best way
to go. Do it with a drive circuit to provide very clean 15V off of a 30V supply
then tack everything ( including all the mixer grounds to that 15V supply.

Yeah, but that virtual ground brings with it it's own set of problems.
For instance, it has to both sink and source current, so you can't just
use a 3 terminal regulator to create the midpoint, although I've seen
schemes with a resistor from virtual ground to negative supply, but
that's not very power efficient - the resistor needs to see, say, 10x
the maximum sink current.

Now that I think about it, some sort of op amp driven appropriately (or
a complementary PNP/NPN pair?) might work - but then you're concerned
about the output Z of the op amp, and how it varies over frequency. And,
of course, the bias current.

We face this in designing low noise instruments for space. You're
running off a DC bus of some sort, and DC/DC inverters tend to be
noisier than straight out buck converters or linear regulators.

On 7/10/22 9:07 AM, Bob kb8tq via time-nuts wrote: > Hi > > Yes it is a pain to implement dual supplies. I ponder that issue every time > I build one of these setups. I’ve built a lot of them …. If you are going to > do a single supply, setting up a “virtual ground” is probably the best way > to go. Do it with a drive circuit to provide very clean 15V off of a 30V supply > then tack everything ( including *all* the mixer grounds to that 15V supply. Yeah, but that virtual ground brings with it it's own set of problems. For instance, it has to both sink and source current, so you can't just use a 3 terminal regulator to create the midpoint, although I've seen schemes with a resistor from virtual ground to negative supply, but that's not very power efficient - the resistor needs to see, say, 10x the maximum sink current. Now that I think about it, some sort of op amp driven appropriately (or a complementary PNP/NPN pair?) might work - but then you're concerned about the output Z of the op amp, and how it varies over frequency. And, of course, the bias current. We face this in designing low noise instruments for space. You're running off a DC bus of some sort, and DC/DC inverters tend to be noisier than straight out buck converters or linear regulators.
G
glenlist
Sun, Jul 10, 2022 11:05 PM

I've ALWAYS used an op-amp to nail down the half rail supply, I never
use the two resistors and cap idea directly  unless the whole thing is
quite basic....

Just don't put bulk capacitance on the output !

like I have seen in some bad designs. ROFL.

Some op-amps are better at this than others- IE dead zones in the
source-sink changeover curve.

On 11/07/2022 3:14 am, Lux, Jim via time-nuts wrote:

On 7/10/22 9:07 AM, Bob kb8tq via time-nuts wrote:

Hi

Yes it is a pain to implement dual supplies. I ponder that issue
every time
I build one of these setups. I’ve

I've ALWAYS used an op-amp to nail down the half rail supply, I never use the two resistors and cap idea directly  unless the whole thing is quite basic.... Just don't put bulk capacitance on the output ! like I have seen in some bad designs. ROFL. Some op-amps are better at this than others- IE dead zones in the source-sink changeover curve. On 11/07/2022 3:14 am, Lux, Jim via time-nuts wrote: > On 7/10/22 9:07 AM, Bob kb8tq via time-nuts wrote: >> Hi >> >> Yes it is a pain to implement dual supplies. I ponder that issue >> every time >> I build one of these setups. I’ve
R(
Richard (Rick) Karlquist
Mon, Jul 11, 2022 1:48 AM

On 7/10/2022 10:14 AM, Lux, Jim via time-nuts wrote:

Yeah, but that virtual ground brings with it it's own set of problems.
For instance, it has to both sink and source current, so you can't just
use a 3 terminal regulator to create the midpoint, although I've seen
schemes with a resistor from virtual ground to negative supply, but
that's not very power efficient - the resistor needs to see, say, 10x
the maximum sink current.

You might look at the LT1118-2.85 "Supply Splitter".  It is able to do
either sink or source, as needed, unlike an ordinary regulator.

Rick N6RK

On 7/10/2022 10:14 AM, Lux, Jim via time-nuts wrote: > > Yeah, but that virtual ground brings with it it's own set of problems. > For instance, it has to both sink and source current, so you can't just > use a 3 terminal regulator to create the midpoint, although I've seen > schemes with a resistor from virtual ground to negative supply, but > that's not very power efficient - the resistor needs to see, say, 10x > the maximum sink current. > You might look at the LT1118-2.85 "Supply Splitter". It is able to do either sink or source, as needed, unlike an ordinary regulator. Rick N6RK
BK
Bob kb8tq
Mon, Jul 11, 2022 3:02 AM

Hi

I seem to recall TI having similar parts.

The big gotcha is tossing large chunks of C onto the ground to rail
connections. A typical op amp is not at all happy with this.

Bob

On Jul 10, 2022, at 5:48 PM, Richard (Rick) Karlquist via time-nuts time-nuts@lists.febo.com wrote:

On 7/10/2022 10:14 AM, Lux, Jim via time-nuts wrote:

Yeah, but that virtual ground brings with it it's own set of problems. For instance, it has to both sink and source current, so you can't just use a 3 terminal regulator to create the midpoint, although I've seen schemes with a resistor from virtual ground to negative supply, but that's not very power efficient - the resistor needs to see, say, 10x the maximum sink current.

You might look at the LT1118-2.85 "Supply Splitter".  It is able to do
either sink or source, as needed, unlike an ordinary regulator.

Rick N6RK


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Hi I seem to recall TI having similar parts. The big gotcha is tossing large chunks of C onto the ground to rail connections. A typical op amp is not at all happy with this. Bob > On Jul 10, 2022, at 5:48 PM, Richard (Rick) Karlquist via time-nuts <time-nuts@lists.febo.com> wrote: > > > > On 7/10/2022 10:14 AM, Lux, Jim via time-nuts wrote: >> Yeah, but that virtual ground brings with it it's own set of problems. For instance, it has to both sink and source current, so you can't just use a 3 terminal regulator to create the midpoint, although I've seen schemes with a resistor from virtual ground to negative supply, but that's not very power efficient - the resistor needs to see, say, 10x the maximum sink current. > > You might look at the LT1118-2.85 "Supply Splitter". It is able to do > either sink or source, as needed, unlike an ordinary regulator. > > Rick N6RK > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com