Discussion of precise voltage measurement

View all threads

Re: [volt-nuts] HP 3457A

CS
Charles Steinmetz
Sun, Aug 11, 2013 12:57 PM

Orrin wrote:

Well, lets discuss the 3456A I got from goldenrubi,
calibrated.  They send a calibration certificate which claims that
their standards are traceable to NIST.  I have no reason to doubt that.

I happened upon one of the seller's auctions, so I checked that one
and several others.  In the boilerplate of each one I found this:

WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE
CALIBRATION CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND
IS ACCEPTED AS WORKING WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.

If I understand this correctly, I buy an instrument, the seller
"calibrates" it before shipment, but doesn't send the calibration
certificate with the item.  He mails it to me after I "accept[] [the
instrument] as working within specs."

So, I have to verify the calibration myself, and then this seller
will send me a "NIST traceable calibration certificate"???  Words
cannot express how irregular that sounds to me.  I'd be interested to
hear about any other NIST traceable calibration facility that works this way.

Best regards,

Charles

Orrin wrote: >Well, lets discuss the 3456A I got from goldenrubi, >calibrated. They send a calibration certificate which claims that >their standards are traceable to NIST. I have no reason to doubt that. I happened upon one of the seller's auctions, so I checked that one and several others. In the boilerplate of each one I found this: >WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE >CALIBRATION CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND >IS ACCEPTED AS WORKING WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE. If I understand this correctly, I buy an instrument, the seller "calibrates" it before shipment, but doesn't send the calibration certificate with the item. He mails it to me after I "accept[] [the instrument] as working within specs." So, I have to verify the calibration myself, and then this seller will send me a "NIST traceable calibration certificate"??? Words cannot express how irregular that sounds to me. I'd be interested to hear about any other NIST traceable calibration facility that works this way. Best regards, Charles
CH
Chuck Harris
Sun, Aug 11, 2013 2:30 PM

I think the statement is awkward English, but what I think he is trying
to say is that:

  1. the instrument will be recalibrated
  2. after the instrument is verified, by the calibration lab,
    to be working within specs,
  3. a NIST traceable calibration certificate will be mailed.

-Chuck Harris

Charles Steinmetz wrote:

Orrin wrote:

Well, lets discuss the 3456A I got from goldenrubi, calibrated.  They send a
calibration certificate which claims that their standards are traceable to NIST.  I
have no reason to doubt that.

I happened upon one of the seller's auctions, so I checked that one and several
others.  In the boilerplate of each one I found this:

WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE CALIBRATION
CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS ACCEPTED AS WORKING
WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.

If I understand this correctly, I buy an instrument, the seller "calibrates" it
before shipment, but doesn't send the calibration certificate with the item.  He
mails it to me after I "accept[] [the instrument] as working within specs."

So, I have to verify the calibration myself, and then this seller will send me a
"NIST traceable calibration certificate"???  Words cannot express how irregular that
sounds to me.  I'd be interested to hear about any other NIST traceable calibration
facility that works this way.

Best regards,

Charles


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I think the statement is awkward English, but what I think he is trying to say is that: 1) the instrument will be recalibrated 2) after the instrument is verified, by the calibration lab, to be working within specs, 3) a NIST traceable calibration certificate will be mailed. -Chuck Harris Charles Steinmetz wrote: > Orrin wrote: > >> Well, lets discuss the 3456A I got from goldenrubi, calibrated. They send a >> calibration certificate which claims that their standards are traceable to NIST. I >> have no reason to doubt that. > > I happened upon one of the seller's auctions, so I checked that one and several > others. In the boilerplate of each one I found this: > >> WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE CALIBRATION >> CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS ACCEPTED AS WORKING >> WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE. > > If I understand this correctly, I buy an instrument, the seller "calibrates" it > before shipment, but doesn't send the calibration certificate with the item. He > mails it to me after I "accept[] [the instrument] as working within specs." > > So, I have to verify the calibration myself, and then this seller will send me a > "NIST traceable calibration certificate"??? Words cannot express how irregular that > sounds to me. I'd be interested to hear about any other NIST traceable calibration > facility that works this way. > > Best regards, > > Charles > > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
CS
Charles Steinmetz
Sun, Aug 11, 2013 4:29 PM

Chuck wrote:

I think the statement is awkward English, but what I think he is
trying to say is that:

  1. the instrument will be recalibrated
  2. after the instrument is verified, by the calibration lab, to be
    working within specs,
  3. a NIST traceable calibration certificate will be mailed.

The seller does his own calibrations.  Also, if he sent it out for
cal, it would come back with a certificate -- why wouldn't he just
leave the certificate in the box when he shipped the unit to the
buyer, instead of mailing it later?  Something very strange seems to
be going on.

Best regards,

Charles

WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE
CALIBRATION
CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS
ACCEPTED AS WORKING
WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.

Chuck wrote: >I think the statement is awkward English, but what I think he is >trying to say is that: >1) the instrument will be recalibrated >2) after the instrument is verified, by the calibration lab, to be >working within specs, >3) a NIST traceable calibration certificate will be mailed. The seller does his own calibrations. Also, if he sent it out for cal, it would come back with a certificate -- why wouldn't he just leave the certificate in the box when he shipped the unit to the buyer, instead of mailing it later? Something very strange seems to be going on. Best regards, Charles >>WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE >>CALIBRATION >>CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS >>ACCEPTED AS WORKING >>WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.
OE
Orin Eman
Sun, Aug 11, 2013 4:50 PM

They shipped the certificate with the 3456A I got from them.  The cal
certificate is from AAA Calibration Equipment Specialist Inc. at a SW
Albuquerque address and is issued to AAA Equipment Resources Inc.
at a NW Albuquerque address.

I suspect some internal accounting reason or a defense against dishonest
buyers for this.

BTW, they don't list any accreditation on the certificate.

I did once ask if they could cal an 8484A and they said no.

As for the local labs here, I'm not happy with the local Tek lab that I
sent my TDS210 to in its original packaging.  They returned it in a small
box with two layers of bubble wrap.

The Fluke HQ and cal lab is twenty miles from here... I think that's about
as good as it gets, so they might be getting my business in the future.

Orin.

On Sun, Aug 11, 2013 at 5:57 AM, Charles Steinmetz csteinmetz@yandex.comwrote:

Orrin wrote:

Well, lets discuss the 3456A I got from goldenrubi, calibrated.  They

send a calibration certificate which claims that their standards are
traceable to NIST.  I have no reason to doubt that.

I happened upon one of the seller's auctions, so I checked that one and
several others.  In the boilerplate of each one I found this:

WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE

CALIBRATION CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS
ACCEPTED AS WORKING WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.

If I understand this correctly, I buy an instrument, the seller
"calibrates" it before shipment, but doesn't send the calibration
certificate with the item.  He mails it to me after I "accept[] [the
instrument] as working within specs."

So, I have to verify the calibration myself, and then this seller will
send me a "NIST traceable calibration certificate"???  Words cannot express
how irregular that sounds to me.  I'd be interested to hear about any other
NIST traceable calibration facility that works this way.

Best regards,

Charles

_____________**
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/**
mailman/listinfo/volt-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

They shipped the certificate with the 3456A I got from them. The cal certificate is from AAA Calibration Equipment Specialist Inc. at a SW Albuquerque address and is issued to AAA Equipment Resources Inc. at a NW Albuquerque address. I suspect some internal accounting reason or a defense against dishonest buyers for this. BTW, they don't list any accreditation on the certificate. I did once ask if they could cal an 8484A and they said no. As for the local labs here, I'm not happy with the local Tek lab that I sent my TDS210 to in its original packaging. They returned it in a small box with two layers of bubble wrap. The Fluke HQ and cal lab is twenty miles from here... I think that's about as good as it gets, so they might be getting my business in the future. Orin. On Sun, Aug 11, 2013 at 5:57 AM, Charles Steinmetz <csteinmetz@yandex.com>wrote: > Orrin wrote: > > Well, lets discuss the 3456A I got from goldenrubi, calibrated. They >> send a calibration certificate which claims that their standards are >> traceable to NIST. I have no reason to doubt that. >> > > I happened upon one of the seller's auctions, so I checked that one and > several others. In the boilerplate of each one I found this: > > WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE >> CALIBRATION CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS >> ACCEPTED AS WORKING WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE. >> > > If I understand this correctly, I buy an instrument, the seller > "calibrates" it before shipment, but doesn't send the calibration > certificate with the item. He mails it to me after I "accept[] [the > instrument] as working within specs." > > So, I have to verify the calibration myself, and then this seller will > send me a "NIST traceable calibration certificate"??? Words cannot express > how irregular that sounds to me. I'd be interested to hear about any other > NIST traceable calibration facility that works this way. > > > Best regards, > > Charles > > > > > ______________________________**_________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/** > mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts> > and follow the instructions there. >
JG
Joseph Gray
Sun, Aug 11, 2013 6:00 PM

Would a 34401A be a better choice, over a 3457A for a budding volt-nut? It
is newer and smaller, and the ebay pricing is similar. Would having that
extra digit via GPIB on the 3457A really be of any practical value?

Joe Gray
W5JG

On Sun, Aug 11, 2013 at 10:29 AM, Charles Steinmetz
csteinmetz@yandex.comwrote:

Chuck wrote:

I think the statement is awkward English, but what I think he is trying

to say is that:

  1. the instrument will be recalibrated
  2. after the instrument is verified, by the calibration lab, to be
    working within specs,
  3. a NIST traceable calibration certificate will be mailed.

The seller does his own calibrations.  Also, if he sent it out for cal, it
would come back with a certificate -- why wouldn't he just leave the
certificate in the box when he shipped the unit to the buyer, instead of
mailing it later?  Something very strange seems to be going on.

Best regards,

Charles

WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE

CALIBRATION
CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS ACCEPTED AS
WORKING
WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.

_____________**
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/**
mailman/listinfo/volt-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Would a 34401A be a better choice, over a 3457A for a budding volt-nut? It is newer and smaller, and the ebay pricing is similar. Would having that extra digit via GPIB on the 3457A really be of any practical value? Joe Gray W5JG On Sun, Aug 11, 2013 at 10:29 AM, Charles Steinmetz <csteinmetz@yandex.com>wrote: > Chuck wrote: > > I think the statement is awkward English, but what I think he is trying >> to say is that: >> 1) the instrument will be recalibrated >> 2) after the instrument is verified, by the calibration lab, to be >> working within specs, >> 3) a NIST traceable calibration certificate will be mailed. >> > > The seller does his own calibrations. Also, if he sent it out for cal, it > would come back with a certificate -- why wouldn't he just leave the > certificate in the box when he shipped the unit to the buyer, instead of > mailing it later? Something very strange seems to be going on. > > Best regards, > > Charles > > > > WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE >>> CALIBRATION >>> CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS ACCEPTED AS >>> WORKING >>> WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE. >>> >> > > > ______________________________**_________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/** > mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts> > and follow the instructions there. > >
MV
Mitch Van Ochten
Sun, Aug 11, 2013 6:51 PM

In 2002 I bought a used HP 34401A.  Sent it in to Agilent (Loveland, CO) for
calibration in March 2005 and they sent it back with both "as found" and "as
left" sets of data.  This is what the "as found" data said:

100mV = -4 ppm
1.0V = +3 ppm
10.0V = +5 ppm
100V = +4 ppm
1000V = +6 ppm

They adjusted it to be perfect.  Sent it again to Agilent in April 2009.
Had to send it to Roseville, CA this time.  (They told me Loveland
calibrations were reserved only for their higher precision meters.) The "as
found" data from Roseville showed the following:

100mV = +7 ppm
1.0V = +1 ppm
10.0V +1 ppm
100V +23 ppm
1000V = +19 ppm

In my opinion the 34401A is an excellent value and this one drifts far less
than it's specified values. Your mileage may vary.

Regards,

mitch

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Joseph Gray
Sent: Sunday, August 11, 2013 2:00 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP 3457A

Would a 34401A be a better choice, over a 3457A for a budding volt-nut? It
is newer and smaller, and the ebay pricing is similar. Would having that
extra digit via GPIB on the 3457A really be of any practical value?

Joe Gray
W5JG

On Sun, Aug 11, 2013 at 10:29 AM, Charles Steinmetz
csteinmetz@yandex.comwrote:

Chuck wrote:

I think the statement is awkward English, but what I think he is
trying

to say is that:

  1. the instrument will be recalibrated
  2. after the instrument is verified, by the calibration lab, to be
    working within specs,
  3. a NIST traceable calibration certificate will be mailed.

The seller does his own calibrations.  Also, if he sent it out for
cal, it would come back with a certificate -- why wouldn't he just
leave the certificate in the box when he shipped the unit to the
buyer, instead of mailing it later?  Something very strange seems to be

going on.

Best regards,

Charles

WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE

CALIBRATION
CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS ACCEPTED
AS WORKING WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.

_____________**
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/**

mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/vol
t-nuts>

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

In 2002 I bought a used HP 34401A. Sent it in to Agilent (Loveland, CO) for calibration in March 2005 and they sent it back with both "as found" and "as left" sets of data. This is what the "as found" data said: 100mV = -4 ppm 1.0V = +3 ppm 10.0V = +5 ppm 100V = +4 ppm 1000V = +6 ppm They adjusted it to be perfect. Sent it again to Agilent in April 2009. Had to send it to Roseville, CA this time. (They told me Loveland calibrations were reserved only for their higher precision meters.) The "as found" data from Roseville showed the following: 100mV = +7 ppm 1.0V = +1 ppm 10.0V +1 ppm 100V +23 ppm 1000V = +19 ppm In my opinion the 34401A is an excellent value and this one drifts far less than it's specified values. Your mileage may vary. Regards, mitch -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Joseph Gray Sent: Sunday, August 11, 2013 2:00 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3457A Would a 34401A be a better choice, over a 3457A for a budding volt-nut? It is newer and smaller, and the ebay pricing is similar. Would having that extra digit via GPIB on the 3457A really be of any practical value? Joe Gray W5JG On Sun, Aug 11, 2013 at 10:29 AM, Charles Steinmetz <csteinmetz@yandex.com>wrote: > Chuck wrote: > > I think the statement is awkward English, but what I think he is > trying >> to say is that: >> 1) the instrument will be recalibrated >> 2) after the instrument is verified, by the calibration lab, to be >> working within specs, >> 3) a NIST traceable calibration certificate will be mailed. >> > > The seller does his own calibrations. Also, if he sent it out for > cal, it would come back with a certificate -- why wouldn't he just > leave the certificate in the box when he shipped the unit to the > buyer, instead of mailing it later? Something very strange seems to be going on. > > Best regards, > > Charles > > > > WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE >>> CALIBRATION >>> CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS ACCEPTED >>> AS WORKING WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE. >>> >> > > > ______________________________**_________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/** > mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/vol t-nuts> > and follow the instructions there. > > _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
M
MARVIN
Mon, Aug 12, 2013 7:21 PM

For USA buyers, Barrytech tracks eBay sold prices for > 10 years and I find his estimates accurate. Despite the site's lack of updates, it compares well to the 3+ years I've tracked weekly prices on various Tek and HP models. I've used his low price estimate as a guide to model's I don't track and had much success.

http://www.barrytech.com/hewlett-packard/meters_digital/hp3457a.html

That said, here's my gestalt of prices; you can decide which model is best.

3455a: $50-100
3456a $50-100
3457a $180-300
34401a $300-700 [ new its $1000].

The 34401a range is particularly wide because its still in production and in catalog, although its days are numbered.

The 3455a does not use the dual slope ADC patented by HP back when, but is now commonplace in many DMMs, so its the least accurate of DMM in the listing.

I would not pay a premium for a "NIST traceable" certificate of an eBay sold device. A good cal is a trust-me event, so impeccable credentials and adherence from the calibrator are needed. Properly done, its easily > 50-100% of low end prices.

----- Original Message -----

From: "Joseph Gray" jgray@zianet.com
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Sunday, August 11, 2013 2:00:29 PM
Subject: Re: [volt-nuts] HP 3457A

Would a 34401A be a better choice, over a 3457A for a budding volt-nut? It
is newer and smaller, and the ebay pricing is similar. Would having that
extra digit via GPIB on the 3457A really be of any practical value?

Joe Gray
W5JG

For USA buyers, Barrytech tracks eBay sold prices for > 10 years and I find his estimates accurate. Despite the site's lack of updates, it compares well to the 3+ years I've tracked weekly prices on various Tek and HP models. I've used his low price estimate as a guide to model's I don't track and had much success. http://www.barrytech.com/hewlett-packard/meters_digital/hp3457a.html That said, here's my gestalt of prices; you can decide which model is best. 3455a: $50-100 3456a $50-100 3457a $180-300 34401a $300-700 [ new its $1000]. The 34401a range is particularly wide because its still in production and in catalog, although its days are numbered. The 3455a does not use the dual slope ADC patented by HP back when, but is now commonplace in many DMMs, so its the least accurate of DMM in the listing. I would not pay a premium for a "NIST traceable" certificate of an eBay sold device. A good cal is a trust-me event, so impeccable credentials and adherence from the calibrator are needed. Properly done, its easily > 50-100% of low end prices. ----- Original Message ----- From: "Joseph Gray" <jgray@zianet.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Sunday, August 11, 2013 2:00:29 PM Subject: Re: [volt-nuts] HP 3457A Would a 34401A be a better choice, over a 3457A for a budding volt-nut? It is newer and smaller, and the ebay pricing is similar. Would having that extra digit via GPIB on the 3457A really be of any practical value? Joe Gray W5JG