volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Re: [volt-nuts] Matched resistors

DM
Dave M
Thu, Jul 24, 2014 4:36 AM

Do you have easy, low cost access to an old Fluke 801 or 803 differential
voltmeter?  These meters have a Kelvin-Varley divider inside that is
composed of strings of resistors that are highly matched in value and
tempco.  If I remember correctly, the highest decade is filled with a string
of 40K resistors, each succeeding smaller decade value is 1/5 of the next
higher decade, but all very well matched. They should all be very well aged
by now too. (Just like wine... they get better with age).
Perhaps you could mix & match values to come up with suitable resistors for
your project.  If your project is a one-off deal, then perhaps that approach
could prove viable.

Cheers,
Dave M

Randy,

Have you considered using multiple identical resistors to reduce the
variance? Depending on who you believe, you can reduce the variance of
the overall resistance by SQRT(N) where N is the number of resistors
in series/parallel. Its not that easy to create a good search query
for
this but here is one such explanation:

http://paulorenato.com/joomla/index.php?option=com_content&view=article&id=109:combining-resistors-to-improve-tolerance&catid=4:projects&Itemid=4

Ideally they should all come from the same batch - ie. manufactured by
the same machine from the same batch of materials. Obviously there's
no
way to guarantee that without close liaison with the manufacturer (you
did want 10 million parts at $.10 each didn't you!) but hopefully a
set
of resistors which come off the same reel would come close.

The absolute value isn't important however, but 'statistical gain'
will also apply to the TCR and stability of the overall divider. The
following assumes that both factors are similarly improved by SQRT(N),
but in fact they may be rather better than that.

That80? or $108 for one sealed Vishay foil divider will buy a lot of
lower spec parts:

Do you have easy, low cost access to an old Fluke 801 or 803 differential voltmeter? These meters have a Kelvin-Varley divider inside that is composed of strings of resistors that are highly matched in value and tempco. If I remember correctly, the highest decade is filled with a string of 40K resistors, each succeeding smaller decade value is 1/5 of the next higher decade, but all very well matched. They should all be very well aged by now too. (Just like wine... they get better with age). Perhaps you could mix & match values to come up with suitable resistors for your project. If your project is a one-off deal, then perhaps that approach could prove viable. Cheers, Dave M > From: Tony <vnuts@toneh.demon.co.uk> > Randy, > > Have you considered using multiple identical resistors to reduce the > variance? Depending on who you believe, you can reduce the variance of > the overall resistance by SQRT(N) where N is the number of resistors > in series/parallel. Its not that easy to create a good search query > for > this but here is one such explanation: > > http://paulorenato.com/joomla/index.php?option=com_content&view=article&id=109:combining-resistors-to-improve-tolerance&catid=4:projects&Itemid=4 > > Ideally they should all come from the same batch - ie. manufactured by > the same machine from the same batch of materials. Obviously there's > no > way to guarantee that without close liaison with the manufacturer (you > did want 10 million parts at $.10 each didn't you!) but hopefully a > set > of resistors which come off the same reel would come close. > > The absolute value isn't important however, but 'statistical gain' > will also apply to the TCR and stability of the overall divider. The > following assumes that both factors are similarly improved by SQRT(N), > but in fact they may be rather better than that. > > That80? or $108 for one sealed Vishay foil divider will buy a lot of > lower spec parts: