time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Re: [time-nuts] SMD TADD-1 distribution amplifier

CS
Charles Steinmetz
Tue, Dec 22, 2015 6:36 PM

Bruce wrote:

Such blanket statements aren't a particularly useful guide unless
calibrated by measurements.


The input voltage noise @1Hz can't be used directly to estimate the
PN noise at 1Hz offset.

My message referred readers to previous list messages for a more
detailed explanation, and recommended a particular search term.  If
you had done that, you would have gone straight to a message in which
I explained that one cannot directly correlate noise density
measurements with PN because each amplifier implementation has its
own specific mechanism of AM-PM conversion.

That said, when one limits the discussion to inductorless, monolithic
amplifiers, I have not observed anywhere near the quantitative
differences between amplifiers (with respect to AM-PM conversion)
that it would take to equalize the PN contributions of two amplifiers
whose baseband noise voltage densities differ by more than 40dB (as
is the case when comparing the MAX477/AD8055 with available low-noise
parts).  So, yes, when a part has an input voltage noise density
that much worse than other available parts, one can conclude
without detailed measurements that it is not a promising candidate
for use as a low-PN amplifier.

Best regards,

Charles

Bruce wrote: >Such blanket statements aren't a particularly useful guide unless >calibrated by measurements. >* * * >The input voltage noise @1Hz can't be used directly to estimate the >PN noise at 1Hz offset. My message referred readers to previous list messages for a more detailed explanation, and recommended a particular search term. If you had done that, you would have gone straight to a message in which I explained that one cannot directly correlate noise density measurements with PN because each amplifier implementation has its own specific mechanism of AM-PM conversion. That said, when one limits the discussion to inductorless, monolithic amplifiers, I have not observed anywhere near the quantitative differences between amplifiers (with respect to AM-PM conversion) that it would take to equalize the PN contributions of two amplifiers whose baseband noise voltage densities differ by more than 40dB (as is the case when comparing the MAX477/AD8055 with available low-noise parts). So, yes, when a part has an input voltage noise density *that* much worse than other available parts, one can conclude without detailed measurements that it is not a promising candidate for use as a low-PN amplifier. Best regards, Charles
BG
Bruce Griffiths
Tue, Dec 22, 2015 11:13 PM

My point was that the PN performance of a device (eg OPA355) with largish
flicker noise can sometimes be a bit better than one might naively expect. The
OPA355's flicker voltage noise is 30dB worse than some wideband bipolar opamps.

Note: I don't recommend this particular fixed gain opamp for critical
applications because its PN floor is relatively high (a bit below -160dBc/Hz
with 10 dBm 10MHz input). I originally intended to use it to buffer the output
of a ringing parallel LC circuit. However a good discrete enhanced CB stage
may also suffice.

The state of the art PN noise for an isolation amp as well as most of the MMIC
devices in Wenzel's amplifier list appears to be around -150dBc/Hz.

It would be useful to have PN plots available for various amplifiers including
some with high flicker noise for comparison purposes.

I can provide some of these.
My instrumentation noise floor (@10MHz) is below -200dBc/Hz if I use the
interferometer.
The corresponding instrument PN floor is around -170dBc/Hz @1Hz (limited
primarily by the splitter PN).
Each measurement typically takes 12 hours or so to achieve this flicker PN
performance.

If I were to find the space for a couple of 10MHz Wilkinson splitters it may be
possible to lower the instrumentation flicker PN to below -170dBc/Hz.

Bruce

On Tuesday, December 22, 2015 01:36:50 PM Charles Steinmetz wrote:

Bruce wrote:

Such blanket statements aren't a particularly useful guide unless
calibrated by measurements.


The input voltage noise @1Hz can't be used directly to estimate the
PN noise at 1Hz offset.

My message referred readers to previous list messages for a more
detailed explanation, and recommended a particular search term.  If
you had done that, you would have gone straight to a message in which
I explained that one cannot directly correlate noise density
measurements with PN because each amplifier implementation has its
own specific mechanism of AM-PM conversion.

That said, when one limits the discussion to inductorless, monolithic
amplifiers, I have not observed anywhere near the quantitative
differences between amplifiers (with respect to AM-PM conversion)
that it would take to equalize the PN contributions of two amplifiers
whose baseband noise voltage densities differ by more than 40dB (as
is the case when comparing the MAX477/AD8055 with available low-noise
parts).  So, yes, when a part has an input voltage noise density
that much worse than other available parts, one can conclude
without detailed measurements that it is not a promising candidate
for use as a low-PN amplifier.

Best regards,

Charles


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the
instructions there.

My point was that the PN performance of a device (eg OPA355) with largish flicker noise can sometimes be a bit better than one might naively expect. The OPA355's flicker voltage noise is 30dB worse than some wideband bipolar opamps. Note: I don't recommend this particular fixed gain opamp for critical applications because its PN floor is relatively high (a bit below -160dBc/Hz with 10 dBm 10MHz input). I originally intended to use it to buffer the output of a ringing parallel LC circuit. However a good discrete enhanced CB stage may also suffice. The state of the art PN noise for an isolation amp as well as most of the MMIC devices in Wenzel's amplifier list appears to be around -150dBc/Hz. It would be useful to have PN plots available for various amplifiers including some with high flicker noise for comparison purposes. I can provide some of these. My instrumentation noise floor (@10MHz) is below -200dBc/Hz if I use the interferometer. The corresponding instrument PN floor is around -170dBc/Hz @1Hz (limited primarily by the splitter PN). Each measurement typically takes 12 hours or so to achieve this flicker PN performance. If I were to find the space for a couple of 10MHz Wilkinson splitters it may be possible to lower the instrumentation flicker PN to below -170dBc/Hz. Bruce On Tuesday, December 22, 2015 01:36:50 PM Charles Steinmetz wrote: > Bruce wrote: > >Such blanket statements aren't a particularly useful guide unless > >calibrated by measurements. > >* * * > >The input voltage noise @1Hz can't be used directly to estimate the > >PN noise at 1Hz offset. > > My message referred readers to previous list messages for a more > detailed explanation, and recommended a particular search term. If > you had done that, you would have gone straight to a message in which > I explained that one cannot directly correlate noise density > measurements with PN because each amplifier implementation has its > own specific mechanism of AM-PM conversion. > > That said, when one limits the discussion to inductorless, monolithic > amplifiers, I have not observed anywhere near the quantitative > differences between amplifiers (with respect to AM-PM conversion) > that it would take to equalize the PN contributions of two amplifiers > whose baseband noise voltage densities differ by more than 40dB (as > is the case when comparing the MAX477/AD8055 with available low-noise > parts). So, yes, when a part has an input voltage noise density > *that* much worse than other available parts, one can conclude > without detailed measurements that it is not a promising candidate > for use as a low-PN amplifier. > > Best regards, > > Charles > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the > instructions there.