Volker,
Silly mistake, I was changing the sample interval in the acquire menu of
TimeLab and it did not match the 5370B. When I changed the display rate
control and used the monitor command to set the sampling interval as
directed by John all was right with the world.
Tom,
What did you do to solve the battery charging circuit issues in the FTS
1050A?
Thanks,
Bob Darby
On 7/8/2013 2:26 PM, Volker Esper wrote:
Bob,
Sorry, I'm not sure, if I've understood the issue - what exactly did
you wrong?
Thank you
Volker
Am 08.07.2013 14:48, schrieb Robert Darby:
John,
After a night's sleep and a rereading of your post I finally
realized what I was doing wrong. I did not understand the the role
of the sampling interval setting and the display rate setting on the
5370. When I follow the process below the results are totally
consistent.
Sorry to taken your time and thanks all for your help. Now to find
the original issue.....
Bob Darby
On 7/7/2013 8:57 PM, John Miles wrote:
So the only difference between the test setups is the setting of the
Display Rate control on the 5370, correct? You're allowing TimeLab
to estimate the sample rate automatically, and giving it enough time
to converge on a stable reading before hitting 'Start Measurement'?
You're correct in that changing the real-world sample rate should
yield results that are identical (or at least very similar) to
resampling the phase data after the fact. In frequency mode, dead
time between readings would make that an iffy proposition, but for
data taken in TI mode the outcomes should be close. -- john, KE5FX
Miles Design LLC
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
No prob! It's a complicated business.
-- john, KE5FX
Miles Design LLC
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Robert Darby
Sent: Monday, July 08, 2013 5:48 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Question about effect of sample interval on ADEV
John,
After a night's sleep and a rereading of your post I finally realized
what I was doing wrong. I did not understand the the role of the
sampling interval setting and the display rate setting on the 5370.
When I follow the process below the results are totally consistent.
Sorry to taken your time and thanks all for your help. Now to find the
original issue.....
Bob Darby
On 7/7/2013 8:57 PM, John Miles wrote:
So the only difference between the test setups is the setting of the
Display Rate control on the 5370, correct? You're allowing TimeLab to
estimate the sample rate automatically, and giving it enough time to
converge on a stable reading before hitting 'Start Measurement'?
You're correct in that changing the real-world sample rate should
yield results that are identical (or at least very similar) to
resampling the phase data after the fact. In frequency mode, dead time
between readings would make that an iffy proposition, but for data
taken in TI mode the outcomes should be close. -- john, KE5FX Miles
Design LLC
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-
nuts
and follow the instructions there.
FYI: I tend to record all serial (RS232/GPIB/USB/LAN) data from counters, analyzers, receivers, environmental sensors with a precision MJD prefix. This allows both tight correlation among different instruments in the lab and also allows ADEV-like tools to estimate, and then gradually refine to high levels of precision, the actual data rate, during data collection. Yes, it adds a few extra bytes, but it can be valuable information sometimes and storage is cheaper than it was a decade ago.
/tvb
----- Original Message -----
From: "John Miles" john@miles.io
To: "'Discussion of precise time and frequency measurement'" time-nuts@febo.com
Sent: Monday, July 08, 2013 2:17 PM
Subject: Re: [time-nuts] Question about effect of sample interval on ADEV
No prob! It's a complicated business.
-- john, KE5FX
Miles Design LLC
-----Original Message-----
From: time-nuts-bounces@febo.com [mailto:time-nuts-bounces@febo.com] On
Behalf Of Robert Darby
Sent: Monday, July 08, 2013 5:48 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Question about effect of sample interval on ADEV
John,
After a night's sleep and a rereading of your post I finally realized
what I was doing wrong. I did not understand the the role of the
sampling interval setting and the display rate setting on the 5370.
When I follow the process below the results are totally consistent.
Sorry to taken your time and thanks all for your help. Now to find the
original issue.....
Bob Darby
On 7/7/2013 8:57 PM, John Miles wrote:
So the only difference between the test setups is the setting of the
Display Rate control on the 5370, correct? You're allowing TimeLab to
estimate the sample rate automatically, and giving it enough time to
converge on a stable reading before hitting 'Start Measurement'?
You're correct in that changing the real-world sample rate should
yield results that are identical (or at least very similar) to
resampling the phase data after the fact. In frequency mode, dead time
between readings would make that an iffy proposition, but for data
taken in TI mode the outcomes should be close. -- john, KE5FX Miles