Do you really have a .01 PPM voltage reference? A personal JJ? That's taking
this hobby to a new level.
Daniel,
They are made like that... Problem is with drift.
When you cal a 3458A the first step is to short the inputs and wait for
the
thermals to die. Then you do a Cal 0 and the incitement stores all the 0
offsets for that set of terminals. they you switch to the other set and do
it
again.
Next step is apply 10 votes to the terminals or in my case it is 9.9999411
and
enter cal 9.9999411 front and back set of terminals. remove the voltage
and
plug in a 10k standard resistor and in my case enter cal 9999.884 for
front and
back. In most cases AC does not have to be done. The meter is comparing
its
measure values with the values entered and calculates the correction
factor
to be used each time a value is displayed.
Most good meters now days do have a null function so you can look at drift
or compare 2 values.
no, this is work and it is not that accurate. that is just our best guess
(about 1 ppm) of an old fluke 732B. We compare with our other 732Bs. Any
available 3458As are monitoring the Flukes. Each hour a program picks up
the Mean Max Min standard deviation from each meter average them together
and plot the Mean for all the meters... Some times I have had 10 meters but
sales have been up so I am down to 1 this week.
I have about 5000 hours of data.
Most 3458As come in within a few ppm of this standard.
We send 3458As out to Agilent for repair and calibration then check our lab
standards.
Old standards and meters seem to be a lot more stable than new ones.
On Wed, Aug 14, 2013 at 9:45 PM, Demian Martin demianm_1@yahoo.com wrote:
Do you really have a .01 PPM voltage reference? A personal JJ? That's
taking
this hobby to a new level.
Daniel,
They are made like that... Problem is with drift.
When you cal a 3458A the first step is to short the inputs and wait for
the
thermals to die. Then you do a Cal 0 and the incitement stores all the 0
offsets for that set of terminals. they you switch to the other set and
do
it
again.
Next step is apply 10 votes to the terminals or in my case it is
9.9999411
and
enter cal 9.9999411 front and back set of terminals. remove the voltage
and
plug in a 10k standard resistor and in my case enter cal 9999.884 for
front and
back. In most cases AC does not have to be done. The meter is comparing
its
measure values with the values entered and calculates the correction
factor
to be used each time a value is displayed.
Most good meters now days do have a null function so you can look at
drift
or compare 2 values.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
--
John Phillips
When I was looking to purchase a 3458A, I spoke to several people including
folks at the Loveland Cal Lab, about whether to purchase 'used' or 'new'.
The message was, as you say, 'old meters seem to be more stable than new' as
explained by the years of 'aging' and letting everything 'stabilize'.
Therefore, I purchased 'used'.
Joe
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of John Phillips
Sent: Thursday, August 15, 2013 1:19 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Agilent calibration
I have about 5000 hours of data.
Most 3458As come in within a few ppm of this standard.
We send 3458As out to Agilent for repair and calibration then check our lab
standards.
Old standards and meters seem to be a lot more stable than new ones.