Randy,
resistor matched in T.C. are extremely expensive, as the manufacturer
(or yourself) would have to select these from a batch of many samples.
reistors with very small T.C. (<1ppm/K) would do the job also, but they
also need to be stable over time, in shelf life opereation mode, i.e.
P<10mW.
That means, you need those hermetically sealed VHP202Z from Vishay, T.C.
is typically < 1ppm/K and they are stable to < 2ppm over 5years. But
they cost already 80€ each, depending on tolerance.
I made a longterm observation of these and found these parameters confirmed.
Frank
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin +/- 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
On Wed, Jul 16, 2014 at 9:40 PM, Frank Stellmach <frank.stellmach@freenet.de
wrote:
Randy,
resistor matched in T.C. are extremely expensive, as the manufacturer (or
yourself) would have to select these from a batch of many samples.
reistors with very small T.C. (<1ppm/K) would do the job also, but they
also need to be stable over time, in shelf life opereation mode, i.e.
P<10mW.
That means, you need those hermetically sealed VHP202Z from Vishay, T.C.
is typically < 1ppm/K and they are stable to < 2ppm over 5years. But they
cost already 80€ each, depending on tolerance.
I made a longterm observation of these and found these parameters
confirmed.
Frank
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
I am not sure if Fluke had matched tempco spare parts, but you can look at
some NOS resistors from Fluke on ebay. They may be a good start and some of
the pricing isn't bad.
Todd
On Thursday, July 17, 2014, Randy Evans wrote:
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin +/- 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
On Wed, Jul 16, 2014 at 9:40 PM, Frank Stellmach <
frank.stellmach@freenet.de javascript:;
wrote:
Randy,
resistor matched in T.C. are extremely expensive, as the manufacturer (or
yourself) would have to select these from a batch of many samples.
reistors with very small T.C. (<1ppm/K) would do the job also, but they
also need to be stable over time, in shelf life opereation mode, i.e.
P<10mW.
That means, you need those hermetically sealed VHP202Z from Vishay, T.C.
is typically < 1ppm/K and they are stable to < 2ppm over 5years. But they
cost already 80€ each, depending on tolerance.
I made a longterm observation of these and found these parameters
confirmed.
Frank
volt-nuts mailing list -- volt-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com javascript:;
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hello Randy,
I am using the LTC1043 in 1/2 VIN or 1/3 VIN configuration.
A matching of the caps is not necessary.
In the 1/2 VIN or 1/3 VIN configuration a matching would give
the advantage that the settling time of the cirquit is reduced.
But in 2* VIN or inverting configuration a matching gives no advantage.
So perhaps it is better to put a 1/2 VIN divider into a feedback loop.
The most important point: you will need a low leakage buffer amplifier
at the output.
The caps should be low leakage foil capacitors. (polypropylene would be
best).
The ESR is negligible against the switch resistance of around 1000 Ohms
And dielectric absorption would also affect only settling time.
In 1/2 VIN configuration I am using cheap small mylar capacitors (WIMA
MKS02)
(isolation time constant is given only with >1250 sec (3000 sec typ)).
Buffer amplifier is a LTC1050.
The circuit is very stable over temperature (10 - 40 deg C).
The absolute amplification error is usually some ppm lower than exact
2:1 value.
(depends somewhat on the pinning which is used so I am not shure wether
the pins are mixed up regarding the charge compensation)
So I dont know wether the ±1 ppm is more a stability figure than a
absolute value.
Even polypropylene capacitors do not change the amplification error.
With best regards
Andreas
Am 17.07.2014 17:26, schrieb Randy Evans:
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An
alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin ± 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
randy,
re. vpg: they offer an hermetic divider (vhd) with typ. 0.1ppm tracking of tempco. if you are ok with lowest available abs. accuracy it will be well priced, I remember having asked them for a price and it was below 50€ for standard lead time (which can be very long). it is also fair to assume it would age more homogeneously than two separate resistors, also you dont need to think about how to keep two resistors at the same temp. make sure each resistor is loaded with less than 10mw, otherwise aging may be above what the claim in their collateral.
Gesendet: Donnerstag, 17. Juli 2014 um 04:40 Uhr
Von: "Frank Stellmach" frank.stellmach@freenet.de
An: volt-nuts@febo.com
Betreff: [volt-nuts] Matched resistors
Randy,
resistor matched in T.C. are extremely expensive, as the manufacturer
(or yourself) would have to select these from a batch of many samples.
reistors with very small T.C. (<1ppm/K) would do the job also, but they
also need to be stable over time, in shelf life opereation mode, i.e.
P<10mW.
That means, you need those hermetically sealed VHP202Z from Vishay, T.C.
is typically < 1ppm/K and they are stable to < 2ppm over 5years. But
they cost already 80€ each, depending on tolerance.
I made a longterm observation of these and found these parameters confirmed.
Frank
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Andreas,
That is good information, I appreciate it. I have contacted LT application
support but they have yet to get back to me on my questions except they did
recommend to use the LTC6943 instead of the LTC1043. Later generation I
guess.
I think i am going to try both the LTC6943 and the LT5400 resistor array
and characterize them. The LT5400 matching ratio looks pretty good over
temp (0.2ppm/C) but the absolute resistor change over temp is -10 to +25
ppm/C, a little larger than I would like for the circuit I am using.
Randy
On Thu, Jul 17, 2014 at 10:37 PM, Andreas Jahn Andreas_-_Jahn@t-online.de
wrote:
Hello Randy,
I am using the LTC1043 in 1/2 VIN or 1/3 VIN configuration.
A matching of the caps is not necessary.
In the 1/2 VIN or 1/3 VIN configuration a matching would give
the advantage that the settling time of the cirquit is reduced.
But in 2* VIN or inverting configuration a matching gives no advantage.
So perhaps it is better to put a 1/2 VIN divider into a feedback loop.
The most important point: you will need a low leakage buffer amplifier at
the output.
The caps should be low leakage foil capacitors. (polypropylene would be
best).
The ESR is negligible against the switch resistance of around 1000 Ohms
And dielectric absorption would also affect only settling time.
In 1/2 VIN configuration I am using cheap small mylar capacitors (WIMA
MKS02)
(isolation time constant is given only with >1250 sec (3000 sec typ)).
Buffer amplifier is a LTC1050.
The circuit is very stable over temperature (10 - 40 deg C).
The absolute amplification error is usually some ppm lower than exact 2:1
value.
(depends somewhat on the pinning which is used so I am not shure wether
the pins are mixed up regarding the charge compensation)
So I dont know wether the ±1 ppm is more a stability figure than a
absolute value.
Even polypropylene capacitors do not change the amplification error.
With best regards
Andreas
Am 17.07.2014 17:26, schrieb Randy Evans:
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin ± 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
I am not familiar with "vpg" or "vhd" terms. I assume the former is a
company (European?) and the latter is a model type? Do you have a URL or
other information to get followup information. The 0.1 ppm tracking is
exactly what I am looking for and I don't need absolute accuracy.
Thanks,
Randy
On Fri, Jul 18, 2014 at 12:00 AM, acbern@gmx.de wrote:
randy,
re. vpg: they offer an hermetic divider (vhd) with typ. 0.1ppm tracking of
tempco. if you are ok with lowest available abs. accuracy it will be well
priced, I remember having asked them for a price and it was below 50€ for
standard lead time (which can be very long). it is also fair to assume it
would age more homogeneously than two separate resistors, also you dont
need to think about how to keep two resistors at the same temp. make sure
each resistor is loaded with less than 10mw, otherwise aging may be above
what the claim in their collateral.
Gesendet: Donnerstag, 17. Juli 2014 um 04:40 Uhr
Von: "Frank Stellmach" frank.stellmach@freenet.de
An: volt-nuts@febo.com
Betreff: [volt-nuts] Matched resistors
Randy,
resistor matched in T.C. are extremely expensive, as the manufacturer
(or yourself) would have to select these from a batch of many samples.
reistors with very small T.C. (<1ppm/K) would do the job also, but they
also need to be stable over time, in shelf life opereation mode, i.e.
P<10mW.
That means, you need those hermetically sealed VHP202Z from Vishay, T.C.
is typically < 1ppm/K and they are stable to < 2ppm over 5years. But
they cost already 80€ each, depending on tolerance.
I made a longterm observation of these and found these parameters
confirmed.
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
I figured out that VPG is Vishay Precision Group but I have not managed to
identify the vhd hermetic divider. Could you point me to the location on
the Vishay web site?
Thanks,
Randy
On Fri, Jul 18, 2014 at 9:00 PM, Randy Evans randyevans2688@gmail.com
wrote:
I am not familiar with "vpg" or "vhd" terms. I assume the former is a
company (European?) and the latter is a model type? Do you have a URL or
other information to get followup information. The 0.1 ppm tracking is
exactly what I am looking for and I don't need absolute accuracy.
Thanks,
Randy
On Fri, Jul 18, 2014 at 12:00 AM, acbern@gmx.de wrote:
randy,
re. vpg: they offer an hermetic divider (vhd) with typ. 0.1ppm tracking
of tempco. if you are ok with lowest available abs. accuracy it will be
well priced, I remember having asked them for a price and it was below 50€
for standard lead time (which can be very long). it is also fair to assume
it would age more homogeneously than two separate resistors, also you dont
need to think about how to keep two resistors at the same temp. make sure
each resistor is loaded with less than 10mw, otherwise aging may be above
what the claim in their collateral.
Gesendet: Donnerstag, 17. Juli 2014 um 04:40 Uhr
Von: "Frank Stellmach" frank.stellmach@freenet.de
An: volt-nuts@febo.com
Betreff: [volt-nuts] Matched resistors
Randy,
resistor matched in T.C. are extremely expensive, as the manufacturer
(or yourself) would have to select these from a batch of many samples.
reistors with very small T.C. (<1ppm/K) would do the job also, but they
also need to be stable over time, in shelf life opereation mode, i.e.
P<10mW.
That means, you need those hermetically sealed VHP202Z from Vishay, T.C.
is typically < 1ppm/K and they are stable to < 2ppm over 5years. But
they cost already 80€ each, depending on tolerance.
I made a longterm observation of these and found these parameters
confirmed.
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
http://www.vishaypg.com/foil-resistors/list/product-63036/
http://www.vishaypg.com/doc?63036
Am 19.07.2014 06:07, schrieb Randy Evans:
I figured out that VPG is Vishay Precision Group but I have not managed to
identify the vhd hermetic divider. Could you point me to the location on
the Vishay web site?
Thanks,
Randy
On Fri, Jul 18, 2014 at 9:00 PM, Randy Evans randyevans2688@gmail.com
wrote:
I am not familiar with "vpg" or "vhd" terms. I assume the former is a
company (European?) and the latter is a model type? Do you have a URL or
other information to get followup information. The 0.1 ppm tracking is
exactly what I am looking for and I don't need absolute accuracy.
Thanks,
Randy
On Fri, Jul 18, 2014 at 12:00 AM, acbern@gmx.de wrote:
randy,
re. vpg: they offer an hermetic divider (vhd) with typ. 0.1ppm tracking
of tempco. if you are ok with lowest available abs. accuracy it will be
well priced, I remember having asked them for a price and it was below 50€
for standard lead time (which can be very long). it is also fair to assume
it would age more homogeneously than two separate resistors, also you dont
need to think about how to keep two resistors at the same temp. make sure
each resistor is loaded with less than 10mw, otherwise aging may be above
what the claim in their collateral.
Gesendet: Donnerstag, 17. Juli 2014 um 04:40 Uhr
Von: "Frank Stellmach" frank.stellmach@freenet.de
An: volt-nuts@febo.com
Betreff: [volt-nuts] Matched resistors
Randy,
resistor matched in T.C. are extremely expensive, as the manufacturer
(or yourself) would have to select these from a batch of many samples.
reistors with very small T.C. (<1ppm/K) would do the job also, but they
also need to be stable over time, in shelf life opereation mode, i.e.
P<10mW.
That means, you need those hermetically sealed VHP202Z from Vishay, T.C.
is typically < 1ppm/K and they are stable to < 2ppm over 5years. But
they cost already 80€ each, depending on tolerance.
I made a longterm observation of these and found these parameters
confirmed.
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hello Randy,
I think the only difference is in oscillator section (and thus power
consumption)
and of cause the TSSOP-package.
The LTC1043 is easily available from stock e.g. from digikey.
The LTC6943 is more difficult to get.
Within the Keithley 2002 LTC1043 is used.
http://www.eevblog.com/forum/testgear/keithley-2002-8-5-digit-dmm-review-and-teardown/
http://dev.xdevs.com/projects/kei2002/repository/entry/photos/K1/small/K2002_1-2251.jpg
I have added a drift chart with longterm drift data.
Note: the drift is for the whole measurement arrangement.
It consists of 3 7V references (2 LTZ1000A and 1 LM399)
measured via a LTC1043 divider with a 24 bit LTC2400 ADC with
temperature compensated voltage reference AD586LQ.
So most of the drift is related to the AD586LQ reference.
(X-axis is in days, Y-axis in ppm)
I get around 2 ppm drift for the LTZ1000A over 1 year
which I guess is mostly humidity related
from the ADC printed cirquit board + AD586 reference drift
and usually below 0.25 ppm standard deviation over 1000 hours. (42 days).
All at unstabilized room temperature.
I guess with resistors you will need ovenized temperature stabilisation
to achieve this.
with best regards
Andreas
Am 19.07.2014 05:57, schrieb Randy Evans:
Andreas,
That is good information, I appreciate it. I have contacted LT application
support but they have yet to get back to me on my questions except they did
recommend to use the LTC6943 instead of the LTC1043. Later generation I
guess.
I think i am going to try both the LTC6943 and the LT5400 resistor array
and characterize them. The LT5400 matching ratio looks pretty good over
temp (0.2ppm/C) but the absolute resistor change over temp is -10 to +25
ppm/C, a little larger than I would like for the circuit I am using.
Randy
On Thu, Jul 17, 2014 at 10:37 PM, Andreas Jahn Andreas_-_Jahn@t-online.de
wrote:
Hello Randy,
I am using the LTC1043 in 1/2 VIN or 1/3 VIN configuration.
A matching of the caps is not necessary.
In the 1/2 VIN or 1/3 VIN configuration a matching would give
the advantage that the settling time of the cirquit is reduced.
But in 2* VIN or inverting configuration a matching gives no advantage.
So perhaps it is better to put a 1/2 VIN divider into a feedback loop.
The most important point: you will need a low leakage buffer amplifier at
the output.
The caps should be low leakage foil capacitors. (polypropylene would be
best).
The ESR is negligible against the switch resistance of around 1000 Ohms
And dielectric absorption would also affect only settling time.
In 1/2 VIN configuration I am using cheap small mylar capacitors (WIMA
MKS02)
(isolation time constant is given only with >1250 sec (3000 sec typ)).
Buffer amplifier is a LTC1050.
The circuit is very stable over temperature (10 - 40 deg C).
The absolute amplification error is usually some ppm lower than exact 2:1
value.
(depends somewhat on the pinning which is used so I am not shure wether
the pins are mixed up regarding the charge compensation)
So I dont know wether the ±1 ppm is more a stability figure than a
absolute value.
Even polypropylene capacitors do not change the amplification error.
With best regards
Andreas
Am 17.07.2014 17:26, schrieb Randy Evans:
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin ± 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Andreas,
Thanks for the information. Do you have the drift chart, etc. posted
anywhere? that would be very interesting reading.
Thanks,
Randy
On Fri, Jul 18, 2014 at 10:11 PM, Andreas Jahn Andreas_-_Jahn@t-online.de
wrote:
Hello Randy,
I think the only difference is in oscillator section (and thus power
consumption)
and of cause the TSSOP-package.
The LTC1043 is easily available from stock e.g. from digikey.
The LTC6943 is more difficult to get.
Within the Keithley 2002 LTC1043 is used.
http://www.eevblog.com/forum/testgear/keithley-2002-8-5-
digit-dmm-review-and-teardown/
http://dev.xdevs.com/projects/kei2002/repository/entry/
photos/K1/small/K2002_1-2251.jpg
I have added a drift chart with longterm drift data.
Note: the drift is for the whole measurement arrangement.
It consists of 3 7V references (2 LTZ1000A and 1 LM399)
measured via a LTC1043 divider with a 24 bit LTC2400 ADC with
temperature compensated voltage reference AD586LQ.
So most of the drift is related to the AD586LQ reference.
(X-axis is in days, Y-axis in ppm)
I get around 2 ppm drift for the LTZ1000A over 1 year
which I guess is mostly humidity related
from the ADC printed cirquit board + AD586 reference drift
and usually below 0.25 ppm standard deviation over 1000 hours. (42 days).
All at unstabilized room temperature.
I guess with resistors you will need ovenized temperature stabilisation to
achieve this.
with best regards
Andreas
Am 19.07.2014 05:57, schrieb Randy Evans:
Andreas,
That is good information, I appreciate it. I have contacted LT
application
support but they have yet to get back to me on my questions except they
did
recommend to use the LTC6943 instead of the LTC1043. Later generation I
guess.
I think i am going to try both the LTC6943 and the LT5400 resistor array
and characterize them. The LT5400 matching ratio looks pretty good over
temp (0.2ppm/C) but the absolute resistor change over temp is -10 to +25
ppm/C, a little larger than I would like for the circuit I am using.
Randy
On Thu, Jul 17, 2014 at 10:37 PM, Andreas Jahn <
Andreas_-_Jahn@t-online.de>
wrote:
Hello Randy,
I am using the LTC1043 in 1/2 VIN or 1/3 VIN configuration.
A matching of the caps is not necessary.
In the 1/2 VIN or 1/3 VIN configuration a matching would give
the advantage that the settling time of the cirquit is reduced.
But in 2* VIN or inverting configuration a matching gives no advantage.
So perhaps it is better to put a 1/2 VIN divider into a feedback loop.
The most important point: you will need a low leakage buffer amplifier at
the output.
The caps should be low leakage foil capacitors. (polypropylene would be
best).
The ESR is negligible against the switch resistance of around 1000 Ohms
And dielectric absorption would also affect only settling time.
In 1/2 VIN configuration I am using cheap small mylar capacitors (WIMA
MKS02)
(isolation time constant is given only with >1250 sec (3000 sec typ)).
Buffer amplifier is a LTC1050.
The circuit is very stable over temperature (10 - 40 deg C).
The absolute amplification error is usually some ppm lower than exact 2:1
value.
(depends somewhat on the pinning which is used so I am not shure wether
the pins are mixed up regarding the charge compensation)
So I dont know wether the ±1 ppm is more a stability figure than a
absolute value.
Even polypropylene capacitors do not change the amplification error.
With best regards
Andreas
Am 17.07.2014 17:26, schrieb Randy Evans:
Frank,
The high cost is my concern, although high performance demands high
price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a
x2
non-inverting amplifier (using a LT1151 precision op amp). An
alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin ± 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps
need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hello Randy,
some information you will get on eevblog. (its much easier to post
(larger) pictures there).
Namely within the LTZ1000, LM399 and T.C. Measurements threads:
http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/
http://www.eevblog.com/forum/projects/lm399-based-10-v-reference/
http://www.eevblog.com/forum/projects/t-c-measurements-on-precision-resistors/
http://www.eevblog.com/forum/projects/oshw-24bit-adc-measurement-system-for-voltage-references/
http://www.eevblog.com/forum/projects/building-a-7-decade-voltage-calibrator/
http://www.eevblog.com/forum/projects/ppmgeek!-5-5-digit-dvm-volt-ref-cal-%28for-arduino-or-any-uc-w-spi%29/msg296127/#msg296127
With best regards
Andreas
Am 19.07.2014 16:33, schrieb Randy Evans:
Andreas,
Thanks for the information. Do you have the drift chart, etc. posted
anywhere? that would be very interesting reading.
Thanks,
Randy
On Fri, Jul 18, 2014 at 10:11 PM, Andreas Jahn Andreas_-_Jahn@t-online.de
wrote:
Hello Randy,
I think the only difference is in oscillator section (and thus power
consumption)
and of cause the TSSOP-package.
The LTC1043 is easily available from stock e.g. from digikey.
The LTC6943 is more difficult to get.
Within the Keithley 2002 LTC1043 is used.
http://www.eevblog.com/forum/testgear/keithley-2002-8-5-
digit-dmm-review-and-teardown/
http://dev.xdevs.com/projects/kei2002/repository/entry/
photos/K1/small/K2002_1-2251.jpg
I have added a drift chart with longterm drift data.
Note: the drift is for the whole measurement arrangement.
It consists of 3 7V references (2 LTZ1000A and 1 LM399)
measured via a LTC1043 divider with a 24 bit LTC2400 ADC with
temperature compensated voltage reference AD586LQ.
So most of the drift is related to the AD586LQ reference.
(X-axis is in days, Y-axis in ppm)
I get around 2 ppm drift for the LTZ1000A over 1 year
which I guess is mostly humidity related
from the ADC printed cirquit board + AD586 reference drift
and usually below 0.25 ppm standard deviation over 1000 hours. (42 days).
All at unstabilized room temperature.
I guess with resistors you will need ovenized temperature stabilisation to
achieve this.
with best regards
Andreas
Am 19.07.2014 05:57, schrieb Randy Evans:
Andreas,
That is good information, I appreciate it. I have contacted LT
application
support but they have yet to get back to me on my questions except they
did
recommend to use the LTC6943 instead of the LTC1043. Later generation I
guess.
I think i am going to try both the LTC6943 and the LT5400 resistor array
and characterize them. The LT5400 matching ratio looks pretty good over
temp (0.2ppm/C) but the absolute resistor change over temp is -10 to +25
ppm/C, a little larger than I would like for the circuit I am using.
Randy
On Thu, Jul 17, 2014 at 10:37 PM, Andreas Jahn <
Andreas_-_Jahn@t-online.de>
wrote:
Hello Randy,
I am using the LTC1043 in 1/2 VIN or 1/3 VIN configuration.
A matching of the caps is not necessary.
In the 1/2 VIN or 1/3 VIN configuration a matching would give
the advantage that the settling time of the cirquit is reduced.
But in 2* VIN or inverting configuration a matching gives no advantage.
So perhaps it is better to put a 1/2 VIN divider into a feedback loop.
The most important point: you will need a low leakage buffer amplifier at
the output.
The caps should be low leakage foil capacitors. (polypropylene would be
best).
The ESR is negligible against the switch resistance of around 1000 Ohms
And dielectric absorption would also affect only settling time.
In 1/2 VIN configuration I am using cheap small mylar capacitors (WIMA
MKS02)
(isolation time constant is given only with >1250 sec (3000 sec typ)).
Buffer amplifier is a LTC1050.
The circuit is very stable over temperature (10 - 40 deg C).
The absolute amplification error is usually some ppm lower than exact 2:1
value.
(depends somewhat on the pinning which is used so I am not shure wether
the pins are mixed up regarding the charge compensation)
So I dont know wether the ±1 ppm is more a stability figure than a
absolute value.
Even polypropylene capacitors do not change the amplification error.
With best regards
Andreas
Am 17.07.2014 17:26, schrieb Randy Evans:
Frank,
The high cost is my concern, although high performance demands high
price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a
x2
non-inverting amplifier (using a LT1151 precision op amp). An
alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin ± 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps
need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Andreas,
Since you are familiar with the LTC1043, do you know what stability over
time and temperature one could expect for the X2 circuit, assuming a high
quality low leakage capacitor was used? Since the circuit does not appear
to be sensitive to the capacitor value, the primary change over time and
temperature would be expected to be the switch resistance I would think.
If the circuit were fed into a high impedance buffer (LTC1151), then I
would expect the switch resistance to have minimal impact. What do you
think?
thanks,
Randy
On Sat, Jul 19, 2014 at 8:58 AM, Andreas Jahn Andreas_-_Jahn@t-online.de
wrote:
Hello Randy,
some information you will get on eevblog. (its much easier to post
(larger) pictures there).
Namely within the LTZ1000, LM399 and T.C. Measurements threads:
http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/
http://www.eevblog.com/forum/projects/lm399-based-10-v-reference/
http://www.eevblog.com/forum/projects/t-c-measurements-on-
precision-resistors/
http://www.eevblog.com/forum/projects/oshw-24bit-adc-
measurement-system-for-voltage-references/
http://www.eevblog.com/forum/projects/building-a-7-decade-
voltage-calibrator/
http://www.eevblog.com/forum/projects/ppmgeek!-5-5-digit-
dvm-volt-ref-cal-%28for-arduino-or-any-uc-w-spi%29/msg296127/#msg296127
With best regards
Andreas
Am 19.07.2014 16:33, schrieb Randy Evans:
Andreas,
Thanks for the information. Do you have the drift chart, etc. posted
anywhere? that would be very interesting reading.
Thanks,
Randy
On Fri, Jul 18, 2014 at 10:11 PM, Andreas Jahn <
Andreas_-_Jahn@t-online.de>
wrote:
Hello Randy,
I think the only difference is in oscillator section (and thus power
consumption)
and of cause the TSSOP-package.
The LTC1043 is easily available from stock e.g. from digikey.
The LTC6943 is more difficult to get.
Within the Keithley 2002 LTC1043 is used.
http://www.eevblog.com/forum/testgear/keithley-2002-8-5-
digit-dmm-review-and-teardown/
http://dev.xdevs.com/projects/kei2002/repository/entry/
photos/K1/small/K2002_1-2251.jpg
I have added a drift chart with longterm drift data.
Note: the drift is for the whole measurement arrangement.
It consists of 3 7V references (2 LTZ1000A and 1 LM399)
measured via a LTC1043 divider with a 24 bit LTC2400 ADC with
temperature compensated voltage reference AD586LQ.
So most of the drift is related to the AD586LQ reference.
(X-axis is in days, Y-axis in ppm)
I get around 2 ppm drift for the LTZ1000A over 1 year
which I guess is mostly humidity related
from the ADC printed cirquit board + AD586 reference drift
and usually below 0.25 ppm standard deviation over 1000 hours. (42
days).
All at unstabilized room temperature.
I guess with resistors you will need ovenized temperature stabilisation
to
achieve this.
with best regards
Andreas
Am 19.07.2014 05:57, schrieb Randy Evans:
Andreas,
That is good information, I appreciate it. I have contacted LT
application
support but they have yet to get back to me on my questions except they
did
recommend to use the LTC6943 instead of the LTC1043. Later generation I
guess.
I think i am going to try both the LTC6943 and the LT5400 resistor array
and characterize them. The LT5400 matching ratio looks pretty good over
temp (0.2ppm/C) but the absolute resistor change over temp is -10 to +25
ppm/C, a little larger than I would like for the circuit I am using.
Randy
On Thu, Jul 17, 2014 at 10:37 PM, Andreas Jahn <
Andreas_-_Jahn@t-online.de>
wrote:
Hello Randy,
I am using the LTC1043 in 1/2 VIN or 1/3 VIN configuration.
A matching of the caps is not necessary.
In the 1/2 VIN or 1/3 VIN configuration a matching would give
the advantage that the settling time of the cirquit is reduced.
But in 2* VIN or inverting configuration a matching gives no advantage.
So perhaps it is better to put a 1/2 VIN divider into a feedback loop.
The most important point: you will need a low leakage buffer amplifier
at
the output.
The caps should be low leakage foil capacitors. (polypropylene would be
best).
The ESR is negligible against the switch resistance of around 1000 Ohms
And dielectric absorption would also affect only settling time.
In 1/2 VIN configuration I am using cheap small mylar capacitors (WIMA
MKS02)
(isolation time constant is given only with >1250 sec (3000 sec typ)).
Buffer amplifier is a LTC1050.
The circuit is very stable over temperature (10 - 40 deg C).
The absolute amplification error is usually some ppm lower than exact
2:1
value.
(depends somewhat on the pinning which is used so I am not shure wether
the pins are mixed up regarding the charge compensation)
So I dont know wether the ±1 ppm is more a stability figure than a
absolute value.
Even polypropylene capacitors do not change the amplification error.
With best regards
Andreas
Am 17.07.2014 17:26, schrieb Randy Evans:
Frank,
The high cost is my concern, although high performance demands high
price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a
x2
non-inverting amplifier (using a LT1151 precision op amp). An
alternative
I am investigating is using the LTC1043 in a voltage doubling circuit
as
shown in Linear Technology app note AN 42, page 6, Figure 16. It
states
that Vout = 2xVin ± 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a
high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on
the
absolute value of the capacitors. I'm not sure if the two 1uF caps
need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
Hello Randy,
I have no experience with a voltage doubler.
From theory the output impedance gets halved by a 2:1 divider
and gets doubled with a X2 multiplier.
So I would expect around a factor 4 more influence of error sources.
But this is only a best guess.
The leakage current of the buffer is usually not the limit.
You have to pay attention on the surface resistance of your PCB
and the leakage of your capacitors.
With best regards
Andreas
Am 21.07.2014 18:38, schrieb Randy Evans:
Andreas,
Since you are familiar with the LTC1043, do you know stability over
time and temperature one could expect for the X2 circuit, assuming a high
quality low leakage capacitor was used? Since the circuit does not appear
to be sensitive to the capacitor value, the primary change over time and
temperature would be expected to be the switch resistance I would think.
If the circuit were fed into a high impedance buffer (LTC1151), then I
would expect the switch resistance to have minimal impact. What do you
think?
thanks,
Randy
On Sat, Jul 19, 2014 at 8:58 AM, Andreas Jahn Andreas_-_Jahn@t-online.de
wrote:
Hello Randy,
some information you will get on eevblog. (its much easier to post
(larger) pictures there).
Namely within the LTZ1000, LM399 and T.C. Measurements threads:
http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/
http://www.eevblog.com/forum/projects/lm399-based-10-v-reference/
http://www.eevblog.com/forum/projects/t-c-measurements-on-
precision-resistors/
http://www.eevblog.com/forum/projects/oshw-24bit-adc-
measurement-system-for-voltage-references/
http://www.eevblog.com/forum/projects/building-a-7-decade-
voltage-calibrator/
http://www.eevblog.com/forum/projects/ppmgeek!-5-5-digit-
dvm-volt-ref-cal-%28for-arduino-or-any-uc-w-spi%29/msg296127/#msg296127
With best regards
Andreas
Am 19.07.2014 16:33, schrieb Randy Evans:
Andreas,
Thanks for the information. Do you have the drift chart, etc. posted
anywhere? that would be very interesting reading.
Thanks,
Randy
On Fri, Jul 18, 2014 at 10:11 PM, Andreas Jahn <
Andreas_-_Jahn@t-online.de>
wrote:
Hello Randy,
I think the only difference is in oscillator section (and thus power
consumption)
and of cause the TSSOP-package.
The LTC1043 is easily available from stock e.g. from digikey.
The LTC6943 is more difficult to get.
Within the Keithley 2002 LTC1043 is used.
http://www.eevblog.com/forum/testgear/keithley-2002-8-5-
digit-dmm-review-and-teardown/
http://dev.xdevs.com/projects/kei2002/repository/entry/
photos/K1/small/K2002_1-2251.jpg
I have added a drift chart with longterm drift data.
Note: the drift is for the whole measurement arrangement.
It consists of 3 7V references (2 LTZ1000A and 1 LM399)
measured via a LTC1043 divider with a 24 bit LTC2400 ADC with
temperature compensated voltage reference AD586LQ.
So most of the drift is related to the AD586LQ reference.
(X-axis is in days, Y-axis in ppm)
I get around 2 ppm drift for the LTZ1000A over 1 year
which I guess is mostly humidity related
from the ADC printed cirquit board + AD586 reference drift
and usually below 0.25 ppm standard deviation over 1000 hours. (42
days).
All at unstabilized room temperature.
I guess with resistors you will need ovenized temperature stabilisation
to
achieve this.
with best regards
Andreas
Am 19.07.2014 05:57, schrieb Randy Evans:
Andreas,
That is good information, I appreciate it. I have contacted LT
application
support but they have yet to get back to me on my questions except they
did
recommend to use the LTC6943 instead of the LTC1043. Later generation I
guess.
I think i am going to try both the LTC6943 and the LT5400 resistor array
and characterize them. The LT5400 matching ratio looks pretty good over
temp (0.2ppm/C) but the absolute resistor change over temp is -10 to +25
ppm/C, a little larger than I would like for the circuit I am using.
Randy
On Thu, Jul 17, 2014 at 10:37 PM, Andreas Jahn <
Andreas_-_Jahn@t-online.de>
wrote:
Hello Randy,
I am using the LTC1043 in 1/2 VIN or 1/3 VIN configuration.
A matching of the caps is not necessary.
In the 1/2 VIN or 1/3 VIN configuration a matching would give
the advantage that the settling time of the cirquit is reduced.
But in 2* VIN or inverting configuration a matching gives no advantage.
So perhaps it is better to put a 1/2 VIN divider into a feedback loop.
The most important point: you will need a low leakage buffer amplifier
at
the output.
The caps should be low leakage foil capacitors. (polypropylene would be
best).
The ESR is negligible against the switch resistance of around 1000 Ohms
And dielectric absorption would also affect only settling time.
In 1/2 VIN configuration I am using cheap small mylar capacitors (WIMA
MKS02)
(isolation time constant is given only with >1250 sec (3000 sec typ)).
Buffer amplifier is a LTC1050.
The circuit is very stable over temperature (10 - 40 deg C).
The absolute amplification error is usually some ppm lower than exact
2:1
value.
(depends somewhat on the pinning which is used so I am not shure wether
the pins are mixed up regarding the charge compensation)
So I dont know wether the ±1 ppm is more a stability figure than a
absolute value.
Even polypropylene capacitors do not change the amplification error.
With best regards
Andreas
Am 17.07.2014 17:26, schrieb Randy Evans:
Frank,
The high cost is my concern, although high performance demands high
price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a
x2
non-inverting amplifier (using a LT1151 precision op amp). An
alternative
I am investigating is using the LTC1043 in a voltage doubling circuit
as
shown in Linear Technology app note AN 42, page 6, Figure 16. It
states
that Vout = 2xVin ± 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a
high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on
the
absolute value of the capacitors. I'm not sure if the two 1uF caps
need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Andreas,
Of course, I would also expect the leakage currents to change over
time/temperature and I would expect them to be the dominant error source.
Randy
On Mon, Jul 21, 2014 at 9:38 AM, Randy Evans randyevans2688@gmail.com
wrote:
Andreas,
Since you are familiar with the LTC1043, do you know what stability over
time and temperature one could expect for the X2 circuit, assuming a high
quality low leakage capacitor was used? Since the circuit does not appear
to be sensitive to the capacitor value, the primary change over time and
temperature would be expected to be the switch resistance I would think.
If the circuit were fed into a high impedance buffer (LTC1151), then I
would expect the switch resistance to have minimal impact. What do you
think?
thanks,
Randy
On Sat, Jul 19, 2014 at 8:58 AM, Andreas Jahn Andreas_-_Jahn@t-online.de
wrote:
Hello Randy,
some information you will get on eevblog. (its much easier to post
(larger) pictures there).
Namely within the LTZ1000, LM399 and T.C. Measurements threads:
http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/
http://www.eevblog.com/forum/projects/lm399-based-10-v-reference/
http://www.eevblog.com/forum/projects/t-c-measurements-on-
precision-resistors/
http://www.eevblog.com/forum/projects/oshw-24bit-adc-
measurement-system-for-voltage-references/
http://www.eevblog.com/forum/projects/building-a-7-decade-
voltage-calibrator/
http://www.eevblog.com/forum/projects/ppmgeek!-5-5-digit-
dvm-volt-ref-cal-%28for-arduino-or-any-uc-w-spi%29/msg296127/#msg296127
With best regards
Andreas
Am 19.07.2014 16:33, schrieb Randy Evans:
Andreas,
Thanks for the information. Do you have the drift chart, etc. posted
anywhere? that would be very interesting reading.
Thanks,
Randy
On Fri, Jul 18, 2014 at 10:11 PM, Andreas Jahn <
Andreas_-_Jahn@t-online.de>
wrote:
Hello Randy,
I think the only difference is in oscillator section (and thus power
consumption)
and of cause the TSSOP-package.
The LTC1043 is easily available from stock e.g. from digikey.
The LTC6943 is more difficult to get.
Within the Keithley 2002 LTC1043 is used.
http://www.eevblog.com/forum/testgear/keithley-2002-8-5-
digit-dmm-review-and-teardown/
http://dev.xdevs.com/projects/kei2002/repository/entry/
photos/K1/small/K2002_1-2251.jpg
I have added a drift chart with longterm drift data.
Note: the drift is for the whole measurement arrangement.
It consists of 3 7V references (2 LTZ1000A and 1 LM399)
measured via a LTC1043 divider with a 24 bit LTC2400 ADC with
temperature compensated voltage reference AD586LQ.
So most of the drift is related to the AD586LQ reference.
(X-axis is in days, Y-axis in ppm)
I get around 2 ppm drift for the LTZ1000A over 1 year
which I guess is mostly humidity related
from the ADC printed cirquit board + AD586 reference drift
and usually below 0.25 ppm standard deviation over 1000 hours. (42
days).
All at unstabilized room temperature.
I guess with resistors you will need ovenized temperature stabilisation
to
achieve this.
with best regards
Andreas
Am 19.07.2014 05:57, schrieb Randy Evans:
Andreas,
That is good information, I appreciate it. I have contacted LT
application
support but they have yet to get back to me on my questions except they
did
recommend to use the LTC6943 instead of the LTC1043. Later generation
I
guess.
I think i am going to try both the LTC6943 and the LT5400 resistor
array
and characterize them. The LT5400 matching ratio looks pretty good
over
temp (0.2ppm/C) but the absolute resistor change over temp is -10 to
+25
ppm/C, a little larger than I would like for the circuit I am using.
Randy
On Thu, Jul 17, 2014 at 10:37 PM, Andreas Jahn <
Andreas_-_Jahn@t-online.de>
wrote:
Hello Randy,
I am using the LTC1043 in 1/2 VIN or 1/3 VIN configuration.
A matching of the caps is not necessary.
In the 1/2 VIN or 1/3 VIN configuration a matching would give
the advantage that the settling time of the cirquit is reduced.
But in 2* VIN or inverting configuration a matching gives no
advantage.
So perhaps it is better to put a 1/2 VIN divider into a feedback loop.
The most important point: you will need a low leakage buffer
amplifier at
the output.
The caps should be low leakage foil capacitors. (polypropylene would
be
best).
The ESR is negligible against the switch resistance of around 1000
Ohms
And dielectric absorption would also affect only settling time.
In 1/2 VIN configuration I am using cheap small mylar capacitors (WIMA
MKS02)
(isolation time constant is given only with >1250 sec (3000 sec
typ)).
Buffer amplifier is a LTC1050.
The circuit is very stable over temperature (10 - 40 deg C).
The absolute amplification error is usually some ppm lower than exact
2:1
value.
(depends somewhat on the pinning which is used so I am not shure
wether
the pins are mixed up regarding the charge compensation)
So I dont know wether the ±1 ppm is more a stability figure than a
absolute value.
Even polypropylene capacitors do not change the amplification error.
With best regards
Andreas
Am 17.07.2014 17:26, schrieb Randy Evans:
Frank,
The high cost is my concern, although high performance demands high
price
typically. I am trying to double the voltage reference from either
an
LM399 or LTZ1000, hence the need for precision matched resistors for
a
x2
non-inverting amplifier (using a LT1151 precision op amp). An
alternative
I am investigating is using the LTC1043 in a voltage doubling
circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It
states
that Vout = 2xVin ± 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a
high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on
the
absolute value of the capacitors. I'm not sure if the two 1uF caps
need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
On 07/17/2014 10:26 AM, Randy Evans wrote:
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin +/- 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Hi Randy,
There are some other error sources that might need to be considered when using
the LTC1043.
I have not used the LTC1043, but note that on the data sheet there is a small
charge injection at each of the switch pins. In the multiply by 2 circuit shown
on the data sheet they are using 1 ufd caps. Typical charge injection (depends
on voltage level) is 8 pC. With the 1 ufd caps this is 8 uV. I assume there is
some offsetting effect - but this might be a significant contributor to the 5
ppm error that is mentioned.
There is also a 6 nA (typical) leakage mentioned. During the "hold" time (
about 1 msec) of the output 1 ufd cap this comes to 6 uV.
Regards,
Bob Smither
yes, charge injection is an issue with all these switches and these also vary, in other words are somewhat unpredictable. now some of this may be compensated by a bigger c, but there are natural limits too. so for a production unit to sell, this would probably be a killer, but also for home-use, how do you predict the behavior over time?
another option to use if you want to stay away from resistors is the pwm-solution as implemented by datron in e.g. their 4910. the pwm signal can today of course be done by uCs.
Gesendet: Montag, 21. Juli 2014 um 22:28 Uhr
Von: "Bob Smither" smither@c-c-i.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] Matched resistors
On 07/17/2014 10:26 AM, Randy Evans wrote:
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin +/- 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Hi Randy,
There are some other error sources that might need to be considered when using
the LTC1043.
I have not used the LTC1043, but note that on the data sheet there is a small
charge injection at each of the switch pins. In the multiply by 2 circuit shown
on the data sheet they are using 1 ufd caps. Typical charge injection (depends
on voltage level) is 8 pC. With the 1 ufd caps this is 8 uV. I assume there is
some offsetting effect - but this might be a significant contributor to the 5
ppm error that is mentioned.
There is also a 6 nA (typical) leakage mentioned. During the "hold" time (
about 1 msec) of the output 1 ufd cap this comes to 6 uV.
Regards,
Bob Smither
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
and yes, I forgot: only down-dividing of course, so to reach 10V, two LTZ1000 would be needed in series. advantage is that noise statistically is reduced by factor of about 1.4. formally also applies to drift.
Gesendet: Montag, 21. Juli 2014 um 20:28 Uhr
Von: "Bob Smither" smither@c-c-i.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] Matched resistors
On 07/17/2014 10:26 AM, Randy Evans wrote:
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin +/- 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Hi Randy,
There are some other error sources that might need to be considered when using
the LTC1043.
I have not used the LTC1043, but note that on the data sheet there is a small
charge injection at each of the switch pins. In the multiply by 2 circuit shown
on the data sheet they are using 1 ufd caps. Typical charge injection (depends
on voltage level) is 8 pC. With the 1 ufd caps this is 8 uV. I assume there is
some offsetting effect - but this might be a significant contributor to the 5
ppm error that is mentioned.
There is also a 6 nA (typical) leakage mentioned. During the "hold" time (
about 1 msec) of the output 1 ufd cap this comes to 6 uV.
Regards,
Bob Smither
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
I agree that there are potentially some serious unknown issues with drift
due to time and temperature due to changes in leakage current, charge
injection, etc. I would think some serious characterization would be
needed before this approach could be used. One approach for the charge
injection is to try and have the input voltage near Vdd/2 so the charge
injection effects are nulled out. I'm trying to figure out how to do that
for a X2 circuit. Any ideas?
I talked to Vishay and they, of course, could do the resistors and I am
awaiting a quote. They stated they would have to "tune" the vhd200
resistor pair to have a very low (0.2ppm/C) ratio stability . Doable but
potentially very expensive. Unfortunately, the maximum value of the
resistors is 20Kohms which would greatly increase the power dissipation of
the resistors, not good for long term stability. Still working the issue.
Randy
On Tue, Jul 22, 2014 at 8:12 AM, acbern@gmx.de wrote:
and yes, I forgot: only down-dividing of course, so to reach 10V, two
LTZ1000 would be needed in series. advantage is that noise statistically is
reduced by factor of about 1.4. formally also applies to drift.
Gesendet: Montag, 21. Juli 2014 um 20:28 Uhr
Von: "Bob Smither" smither@c-c-i.com
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] Matched resistors
On 07/17/2014 10:26 AM, Randy Evans wrote:
Frank,
The high cost is my concern, although high performance demands high
price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a
x2
non-inverting amplifier (using a LT1151 precision op amp). An
alternative
I am investigating is using the LTC1043 in a voltage doubling circuit
as
shown in Linear Technology app note AN 42, page 6, Figure 16. It
states
that Vout = 2xVin +/- 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on
the
absolute value of the capacitors. I'm not sure if the two 1uF caps
need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Hi Randy,
There are some other error sources that might need to be considered when
using
the LTC1043.
I have not used the LTC1043, but note that on the data sheet there is a
small
charge injection at each of the switch pins. In the multiply by 2
circuit shown
on the data sheet they are using 1 ufd caps. Typical charge injection
(depends
on voltage level) is 8 pC. With the 1 ufd caps this is 8 uV. I assume
there is
some offsetting effect - but this might be a significant contributor to
the 5
ppm error that is mentioned.
There is also a 6 nA (typical) leakage mentioned. During the "hold"
time (
about 1 msec) of the output 1 ufd cap this comes to 6 uV.
Regards,
Bob Smither
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Randy,
Have you considered using multiple identical resistors to reduce the
variance? Depending on who you believe, you can reduce the variance of
the overall resistance by SQRT(N) where N is the number of resistors in
series/parallel. Its not that easy to create a good search query for
this but here is one such explanation:
Ideally they should all come from the same batch - ie. manufactured by
the same machine from the same batch of materials. Obviously there's no
way to guarantee that without close liaison with the manufacturer (you
did want 10 million parts at $.10 each didn't you!) but hopefully a set
of resistors which come off the same reel would come close.
The absolute value isn't important however, but 'statistical gain' will
also apply to the TCR and stability of the overall divider. The
following assumes that both factors are similarly improved by SQRT(N),
but in fact they may be rather better than that.
That80€ or $108 for one sealed Vishay foil divider will buy a lot of
lower spec parts:
Approx 12558 x Susumu RR0510P .5%, 25ppm 0402 (Digikey, $86/10k). 6279
in series and parallel in each leg of the 1:1
dividerhttp://media.digikey.com/photos/Susumu%20Photos/RR%200402%20SERIES.jpg
might reduce the variance to 25ppm/SQRT(6279) = .32ppm. Can't see any
spec for stability, but it may also improve similarly. Would take a
while to solder them onto stripboard though!
Slightly more sensible might be 1078 x TE Connectivity RP73 1%, 10ppm
1206 (Digikey, $100.18/1K). Stability .5% (no qualifers in datasheet)
=> 10ppm/SQRT(539) = .43ppm, stability => 215ppm
Or 372 x KOA Speer RN731JTTD4021B5 .1%, 5ppm (Mouser, $29/100).
Stability not on data sheet but typical endurance is +/- .02% for 1000
hrs @ 70C on/off 1.5hours/.5hours.
=> 5ppm/SQRT(138) = .37ppm, endurance => 14.7ppm (Stability should be
rather better than that). Note that the Mouser part no. is for a 25ppm
part but their manufacturer's part number is the 5ppm part as is the
description. Also, the price is way too high for 25ppm parts.
Or 28 x Susumu RG2012L .01%, 2ppm (Digikey, $39.6/10). Stability not
quoted but typical Load Life is .01% (1000 x 1.5hours on/.5hours off at 85C)
=> 2ppm/SQRT(14) = .53ppm, endurance => 27ppm
You could also use multiple resistor networks. Eg:
104 x Susumu RM2012B-103/103-PBVW10 .1%, 5ppm tracking, 2
resistors/device (Digikey $104/100). Stability not quoted, endurance
500ppm (1000 x 1.5hours on/.5hours off at 85C)
=> 5ppm/SQRT(104) = .49ppm, endurance => 49ppm
35 x TT Electronics SFN08B4701CBQLF7, .25%, 5ppm tracking 7
resistors/device (Digikey, $76/25) . Stability not quoted, high
temperature exposure < 1000ppm
=> 5ppm/SQRT(122) = .52ppm
33 x TT Electronics 668A1001DLF .5%, 5ppm tracking 8resistors/device
(Digikey, $82/25). Stability not quoted, load life < 1000ppm
=> 5ppm/SQRT(33 * 4) = .45ppm
16 x Vishay DFN .1%, 3ppm tracking with 4 resistors/device (Digikey,
$5.24/1). Shelf life ratio stability is specced at 20ppm (1 year at
25C). (That may be a typical rather than a maximum - your parts may all
be much worse than typical). The 3ppm tracking TCR may also be a typical
figure as its headlined in a section titled 'TYPICAL PERFORMANCE' but in
the specification table its not qualified with '(typical)' as they
sometimes do in other datasheets. Its hard to tell.
=> 3ppm/SQRT(32) = .53ppm shelf life stability => 3.5ppm
5 x Vishay DSMZ metal foil dividers, .5ppm tracking max (probably
performs rather better than this over restricted temperature range, but
don't believe the Vishay typical figure of < .1ppm/C) (Digikey,
$22.93/1). Shelf life ratio stability not quoted but 'typical limit' for
Load Life ratio stability is 50ppm (2000 hours at 70C). Who knows what a
typical limit is? Again, probably best to treat Vishay 'typical' figures
with a pinch of salt given the experience of another poster on volt-nuts.
=> .5ppm/SQRT(5) = .22ppm, load life => 22ppm
Interestingly Digikey quote a price of only $5400 for 1k parts for the
similar DSM divider (1ppm tracking), which is a huge difference from
$22.93. Might be worth considering a bulk buy if there enough volt-nuts
with the same problem. They aren't stocked though so that price might
not be 'real'. However:
20 x Vishay DSM dividers, 1ppm (Digikey, $5400/1000) Load life ratio
stability 'typical limit' 50ppm
=> 1ppm/SQRT(20) = .22ppm, load life => 11ppm
Multiple LT5400 networks could also be used and may give the best
results, but the much larger absolute tolerance, +/-15% would cause
those with the highest value for series connected/lowest for parallel to
dominate and reduce the statistical improvement. Do your own calculations.
Its interesting that all these different components end up providing
pretty much the same performance for the same cost - in other words the
cost is inversely proportional to the TCR^2
My gut feeling is that the tracking TCR will improve rather better than
the SQRT(N) calculated, if they do indeed come from the same batch, as I
would expect them to have similar absolute TCRs. Thus you might be able
to get away with rather less parts to achieve < 1ppm. The SQRT(N) factor
comes from assuming that the variation in the value is random, and I
believe, has a particular distribution (Guassian or normal?). Component
specifications are often derived from the distribution parameters
measured from a large set of production samples, with the max/min values
determined from a multiple (typically 6?) of the standard deviations of
the distribution? The worst case specifications for TCR and stability
may (I don't know, just hypothesizing) be derived very differently. For
example, the TCR may be affected not only by the characteristics of the
bulk resistive material, but also due to stresses on the element due to
thermal expansion of the substrate/packaging. It may be that the former
is almost identical for all components from the batch, but the latter is
less predictable. The specification max/min would have to allow for the
worst cases which might be due to a relatively few which for some
reason (microcracking in the substrate perhaps) have much larger
variance from the majority. The distribution of TCRs from a set of
resistors could be very skewed with long tails and the SQRT(N) reduction
in variance may be well off the mark.
Stability is more difficult because the shelf life stability is rarely
specified, but is likely to be the closest to your usage. For reference,
the Vishay DFSMZ datasheet specifies ratio stability of .015% for 2000
hour at 70C and .002% for shelf life ratio stability. The 7.5X
difference might be useful for estimating shelf life stability for
resistors that only quote load life or endurance specs. But it might
not! I'm not sure that the endurance spec is very useful either as it
subjects the resistor to a large number of large temperature cycles
which won't be anywhere near your usage.
I would expect the long term tracking stability to be much better than
(worst case datasheet stability)/SQRT(N) as I would expect the vast
majority to age in similar ways, if not by the same magnitude. Whilst
the specs show stability to be +/- xx% I would expect that most will age
in the same way - probably slowly increasing resistance over time. I
also expect there are experienced posters here who know otherwise!
Similarly to TCR, it could be that for example, the stability of most
resistors in a batch may be quite good, but the specs reflect that a few
may be much worse due to random faults in individual samples - such as
defects in the protective coating of the element allowing corrosion to
occur in a few samples. You'd need a very good understanding of the
factors that determine the resistor stability to calculate the overall
stability of multiple resistors.
I would expect similar factors to apply to ratio tracking due to
humidity changes. No doubt there is some useful information out their in
application notes/research papers on the variance in long term stability
between resistors of various types (and maybe even for parts taken from
the same batch) just waiting for some interested volt-nut to discover?
The fewer the parts, the more chance of statistical outliers reducing
the improvement over a single part, but you could test each divider for
the best matching, if you've got a decent meter, fairly easily by
applying a voltage from a stable, low noise source (a battery would be
good if its temperature is kept very stable), and measure the voltage at
the centre tap. Then put the resistor network in a plastic bag and
immerse it in boiling water to raise the temperature by 75C or so; .5ppm
tracking would give 9.4uV/V maximum change; you'd probably need to
reverse the meter leads a few times to null out thermal EMFs.
Alternatively measure the voltage difference between the divider under
test and another driven by the same voltage source and kept at a stable
temperature - ie. in a bridge configuration. A simple high gain
amplifier (say 1000x) with adjustable offset would allow testing with a
more realistic lower temperature difference of say 20C and/or a cheap meter.
Accuracy is not particularly important - you probably don't need to know
the temperature tracking coefficient to better than 20%.
Component layout would need to ensure any thermal gradients apply
equally to both legs of the divider by interleaving upper and lower
resistors.
Tony H
On 17/07/2014 16:26, Randy Evans wrote:
Frank,
The high cost is my concern, although high performance demands high price
typically. I am trying to double the voltage reference from either an
LM399 or LTZ1000, hence the need for precision matched resistors for a x2
non-inverting amplifier (using a LT1151 precision op amp). An alternative
I am investigating is using the LTC1043 in a voltage doubling circuit as
shown in Linear Technology app note AN 42, page 6, Figure 16. It states
that Vout = 2xVin +/- 5 ppm. I am less concerned about the absolute
accuracy than I am about the long term stability. I assume that a high
quality capacitor is required (low leakage, low ESR, low dielectric
absorbtion, etc.) but the circuit does not appear to be dependent on the
absolute value of the capacitors. I'm not sure if the two 1uF caps need
to be matched. If they do then that would be a show stopper.
Does anyone have any experience using the LTC1043 in such a circuit?
Thanks,
Randy
On Wed, Jul 16, 2014 at 9:40 PM, Frank Stellmach <frank.stellmach@freenet.de
wrote:
Randy,
resistor matched in T.C. are extremely expensive, as the manufacturer (or
yourself) would have to select these from a batch of many samples.
reistors with very small T.C. (<1ppm/K) would do the job also, but they
also need to be stable over time, in shelf life opereation mode, i.e.
P<10mW.
That means, you need those hermetically sealed VHP202Z from Vishay, T.C.
is typically < 1ppm/K and they are stable to < 2ppm over 5years. But they
cost already 80€ each, depending on tolerance.
I made a longterm observation of these and found these parameters
confirmed.
Frank
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.