Re: [volt-nuts] vSolartron7081 rom reading (m k)

BE
Bill Ezell
Fri, Jan 18, 2013 1:20 AM

BTW, make absolutely sure you save the Zener current token value
somewhere safe. The rest of the cal constants can be recovered with some
good calibration standards, but the Zener value takes some pretty
specialized calibration.

I think the performance of the 7081 is really phenomenal given that the
reference is a non-ovenized Zener. It gets this by individually setting
the Zener current to the value that gives the minimal voltage change for
temperature change, then injecting a 2nd-order correction factor.

The moral of this story is that the current token setting is specific to
the Zener, and you'd need a good temperature-controlled chamber and a
lot of time to replicate.

--
Bill Ezell

They said 'Windows or better'
so I used Linux.

BTW, make absolutely sure you save the Zener current token value somewhere safe. The rest of the cal constants can be recovered with some good calibration standards, but the Zener value takes some pretty specialized calibration. I think the performance of the 7081 is really phenomenal given that the reference is a non-ovenized Zener. It gets this by individually setting the Zener current to the value that gives the minimal voltage change for temperature change, then injecting a 2nd-order correction factor. The moral of this story is that the current token setting is specific to the Zener, and you'd need a good temperature-controlled chamber and a lot of time to replicate. -- Bill Ezell ---------- They said 'Windows or better' so I used Linux.
DC
David C. Partridge
Fri, Jan 18, 2013 1:30 PM

The zener token value is written on the lower right PCB in the rear right hand corner.

Dave
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Bill Ezell
Sent: 18 January 2013 01:21
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] vSolartron7081 rom reading (m k)

BTW, make absolutely sure you save the Zener current token value somewhere safe.

The zener token value is written on the lower right PCB in the rear right hand corner. Dave -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Bill Ezell Sent: 18 January 2013 01:21 To: volt-nuts@febo.com Subject: Re: [volt-nuts] vSolartron7081 rom reading (m k) BTW, make absolutely sure you save the Zener current token value somewhere safe.